Christof Schwiening |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, I wonder if someone could throw some light on a problem we have been having for the past few years! We use a Leica SP5 confocal on an upright microscope with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During solution changes - most notably the addition and removal of organic compounds such as propionate (20 mM)- we get severe image distortion and changes in focal depth. I attempted to attach a graph of a ROI from a fixed sealed specimen which we have imaged during solution changes with 3 different objectives - however, the ListServer rejected every image format I tried...I can send it by email if anyone is interested. It is impossible that the propionate is actually getting to the specimen itself - I suspect it must be as a result of refractive index changes. My questions are: How do others image small structures (i.e. dendritic spines or NMJs) during changes of solution with differing refractive indicies? Is there some well know trick? Or, am I using the wrong objectives? Does wide field microscopy have the same problems? Many thanks, Christof |
![]() ![]() |
Vladimir Ghukasyan-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Christof, As you mentioned, one of the most prominent reasons for the distortions you see is change of the refractive index. Lenses are usually corrected for chromatic and spherical aberrations for some defined refractive index. The more you deviate from it, the more will be the extent of the aberrations affecting your image. One way of overcoming this is using lenses with adjustable refractive index correction collar. With regards, Vladimir On Mon, Mar 18, 2013 at 11:29 AM, Christof Schwiening <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > I wonder if someone could throw some light on a problem we have been having > for the past few years! We use a Leica SP5 confocal on an upright microscope > with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During > solution changes - most notably the addition and removal of organic > compounds such as propionate (20 mM)- we get severe image distortion and > changes in focal depth. I attempted to attach a graph of a ROI from a fixed > sealed specimen which we have imaged during solution changes with 3 > different objectives - however, the ListServer rejected every image format I > tried...I can send it by email if anyone is interested. It is impossible that the > propionate is actually getting to the specimen itself - I suspect it must be as a > result of refractive index changes. > > My questions are: How do others image small structures (i.e. dendritic spines > or NMJs) during changes of solution with differing refractive indicies? Is there > some well know trick? Or, am I using the wrong objectives? Does wide field > microscopy have the same problems? > Many thanks, > Christof |
![]() ![]() |
Julio Vazquez |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Christof, I don't actually have an answer to your problem. However, my feeling is that going from let's say water to 20 mM propionate should have only a very minimal effect on refractive index (RI water = 1.333, 100% propionate: 1.386; 20 mM propionate maybe ~ 1.335; check tables/charts listing RI as a function of concentration for propionate or another similar substance). Adjusting the refractive index collar as suggested by Vladimir is one good idea, if you have one; those however are found most likely on high NA immersion objectives, not so much on dipping objectives. I guess what may be worth knowing is which specific objectives you are using, and also what the distortions look like. Is it possible that you are getting your cells dry and wet again (in which case maybe the cells themselves may be doing something funky); or they may not like to be in propionic acid; or is it possible you are trapping air bubbles under your objective as you replace medium? Julio Vazquez, Fred Hutchinson Cancer Research Center http://www.fhcrc.org == On Mar 18, 2013, at 9:24 AM, Vladimir Ghukasyan wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello Christof, > > As you mentioned, one of the most prominent reasons for the > distortions you see is change of the refractive index. Lenses are > usually corrected for chromatic and spherical aberrations for some > defined refractive index. The more you deviate from it, the more will > be the extent of the aberrations affecting your image. One way of > overcoming this is using lenses with adjustable refractive index > correction collar. > > With regards, > Vladimir > > On Mon, Mar 18, 2013 at 11:29 AM, Christof Schwiening <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear All, >> I wonder if someone could throw some light on a problem we have been having >> for the past few years! We use a Leica SP5 confocal on an upright microscope >> with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During >> solution changes - most notably the addition and removal of organic >> compounds such as propionate (20 mM)- we get severe image distortion and >> changes in focal depth. I attempted to attach a graph of a ROI from a fixed >> sealed specimen which we have imaged during solution changes with 3 >> different objectives - however, the ListServer rejected every image format I >> tried...I can send it by email if anyone is interested. It is impossible that the >> propionate is actually getting to the specimen itself - I suspect it must be as a >> result of refractive index changes. >> >> My questions are: How do others image small structures (i.e. dendritic spines >> or NMJs) during changes of solution with differing refractive indicies? Is there >> some well know trick? Or, am I using the wrong objectives? Does wide field >> microscopy have the same problems? >> Many thanks, >> Christof |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The list doesn't allow image attachments as everyone's inbox would quickly fill up. Your best bet is to host the image on dropbox or the like and include the link to the image in your email. Craig On Mon, Mar 18, 2013 at 9:29 AM, Christof Schwiening <[hidden email]>wrote: > ***** > I attempted to attach a graph of a ROI from a fixed > sealed specimen which we have imaged during solution changes with 3 > different objectives - however, the ListServer rejected every image format > I > tried...I can send it by email if anyone is interested. Christof > |
![]() ![]() |
Arvydas Matiukas |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Christof, More quantitative description of your imaging geometry/optics would be useful. I have been observing similar effects when imaging a thick sample on inverted scope using 40x water dipping lens in immersion mode. When water drop substantially shrinks after 2-3 hrs focus starts shifting until disappears completely (from the set range). This corresponds to ultimate refractive index change from 1.333 to 1.0. The possible solution is to set big enough thickness of z-stack so that your specimen still fits into imaged focal range (taking into account the z-range shift due to refractive index change). I would like to see your graph and the distorted image. I wonder if your chemicals are not dissolving well or recrystalizing thus creating nonhomogenous layers distorting your image. Or maybe some vibrations create inhomogeneities in the immersion/dipping medium. Do you observe this type of distortion with fixed specimens as well? Best regards, Arvydas -------------------------------- Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core Department of Neurosci& Physiology SUNY Upstate Medical University 766 Irving Ave., WH 3167 Syracuse, NY 13210 tel.: 315-464-7997 fax: 315-464-8014 email: [hidden email] >>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, I wonder if someone could throw some light on a problem we have been having for the past few years! We use a Leica SP5 confocal on an upright microscope with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During solution changes - most notably the addition and removal of organic compounds such as propionate (20 mM)- we get severe image distortion and changes in focal depth. I attempted to attach a graph of a ROI from a fixed sealed specimen which we have imaged during solution changes with 3 different objectives - however, the ListServer rejected every image format I tried...I can send it by email if anyone is interested. It is impossible that the propionate is actually getting to the specimen itself - I suspect it must be as a result of refractive index changes. My questions are: How do others image small structures (i.e. dendritic spines or NMJs) during changes of solution with differing refractive indicies? Is there some well know trick? Or, am I using the wrong objectives? Does wide field microscopy have the same problems? Many thanks, Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** On Mon, 18 Mar 2013 10:13:51 -0700, Julio Vazquez <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** Dear Julio, There is no adjustable collar on the objectives - the link below shows the list of dipping objectives I have used and what a ROI plot looks like. https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg The collar adjustment would be of little use since I need to record changes during solution switches. My cells are definitely always staying wet. The effect is nothing to do with the effect of propionate on the cells since I see the distortions when I run the solution over a fixed glass slide with a preparation beneath a coverslip (embeded and sealed). I agree that the refractive index change is small... There are no air bubbles beneath the objective. I think it must be a Schlieren type of effect. Greetings, Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Craig, Here is the link: https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg Greetings, Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Vladimir, Yes, but I need to record during the solution changes - on a second by second basis - and there is no correction collar.... Greetings, Christof On Mon, 18 Mar 2013 12:24:01 -0400, Vladimir Ghukasyan <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello Christof, > >As you mentioned, one of the most prominent reasons for the >distortions you see is change of the refractive index. Lenses are >usually corrected for chromatic and spherical aberrations for some >defined refractive index. The more you deviate from it, the more will >be the extent of the aberrations affecting your image. One way of >overcoming this is using lenses with adjustable refractive index >correction collar. > >With regards, >Vladimir > >On Mon, Mar 18, 2013 at 11:29 AM, Christof Schwiening <[hidden email]> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear All, >> I wonder if someone could throw some light on a problem we have been having >> for the past few years! We use a Leica SP5 confocal on an upright microscope >> with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During >> solution changes - most notably the addition and removal of organic >> compounds such as propionate (20 mM)- we get severe image distortion and >> changes in focal depth. I attempted to attach a graph of a ROI from a fixed >> sealed specimen which we have imaged during solution changes with 3 >> different objectives - however, the ListServer rejected every image format I >> tried...I can send it by email if anyone is interested. It is impossible that the >> propionate is actually getting to the specimen itself - I suspect it must be as a >> result of refractive index changes. >> >> My questions are: How do others image small structures (i.e. dendritic spines >> or NMJs) during changes of solution with differing refractive indicies? Is there >> some well know trick? Or, am I using the wrong objectives? Does wide field >> microscopy have the same problems? >> Many thanks, >> Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Arvydas, The graph is here: https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg I have tried three long working distance objectives (20-63X). I don't have a good example of the distorted image since it shifts faster than my resonnant scanner can capture it. However, the ROI plot shows the magnitude of the shifts on the fluorescence signal. We do see the same effect on fixed samples (which is what the image above shows). I am sure that it is some kind of Schlieren type of effect (refractive index). I think that in areas of high contrast even small optical deviations cause quite marked intensity changes..... Greetings, Christof On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi Christof, > >More quantitative description of your imaging geometry/optics >would be useful. I have been observing similar effects when imaging >a thick sample on >inverted scope using 40x water dipping lens in immersion mode. >When water drop substantially shrinks after 2-3 hrs >focus starts shifting until disappears completely (from the set range). >This corresponds to ultimate refractive index change from 1.333 to 1.0. >The possible solution is to set big enough thickness of z-stack so that >your specimen still fits into imaged focal range (taking into account the >z-range shift due to refractive index change). > >I would like to see your graph and the distorted image. I wonder >if your chemicals are not dissolving well or recrystalizing thus creating >nonhomogenous layers distorting your image. Or maybe some vibrations >create inhomogeneities in the immersion/dipping medium. Do you observe this >type of distortion with fixed specimens as well? > >Best regards, >Arvydas >-------------------------------- > > > > >Arvydas Matiukas, Ph.D. >Director of Confocal&Two-Photon Core >Department of Neurosci& Physiology >SUNY Upstate Medical University >766 Irving Ave., WH 3167 >Syracuse, NY 13210 >tel.: 315-464-7997 >fax: 315-464-8014 >email: [hidden email] >>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear All, >I wonder if someone could throw some light on a problem we have been >for the past few years! We use a Leica SP5 confocal on an upright microscope >with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During >solution changes - most notably the addition and removal of organic >compounds such as propionate (20 mM)- we get severe image distortion and >changes in focal depth. I attempted to attach a graph of a ROI from a fixed >sealed specimen which we have imaged during solution changes with 3 >different objectives - however, the ListServer rejected every image format I >tried...I can send it by email if anyone is interested. It is impossible that the >propionate is actually getting to the specimen itself - I suspect it must be as a >result of refractive index changes. > >My questions are: How do others image small structures (i.e. dendritic spines >or NMJs) during changes of solution with differing refractive indicies? Is there >some well know trick? Or, am I using the wrong objectives? Does wide field >microscopy have the same problems? >Many thanks, >Christof |
![]() ![]() |
Mark Cannell-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** It looks like a movement problem to me… Cheers On 19/03/2013, at 3:02 PM, Christof Schwiening <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Arvydas, > The graph is here: https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg > I have tried three long working distance objectives (20-63X). I don't have a > good example of the distorted image since it shifts faster than my resonnant > scanner can capture it. However, the ROI plot shows the magnitude of the > shifts on the fluorescence signal. We do see the same effect on fixed samples > (which is what the image above shows). I am sure that it is some kind of > Schlieren type of effect (refractive index). I think that in areas of high > contrast even small optical deviations cause quite marked intensity changes..... > Greetings, > Christof > > On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas > <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi Christof, >> >> More quantitative description of your imaging geometry/optics >> would be useful. I have been observing similar effects when imaging >> a thick sample on >> inverted scope using 40x water dipping lens in immersion mode. >> When water drop substantially shrinks after 2-3 hrs >> focus starts shifting until disappears completely (from the set range). >> This corresponds to ultimate refractive index change from 1.333 to 1.0. >> The possible solution is to set big enough thickness of z-stack so that >> your specimen still fits into imaged focal range (taking into account the >> z-range shift due to refractive index change). >> >> I would like to see your graph and the distorted image. I wonder >> if your chemicals are not dissolving well or recrystalizing thus creating >> nonhomogenous layers distorting your image. Or maybe some vibrations >> create inhomogeneities in the immersion/dipping medium. Do you observe this >> type of distortion with fixed specimens as well? >> >> Best regards, >> Arvydas >> -------------------------------- >> >> >> >> >> Arvydas Matiukas, Ph.D. >> Director of Confocal&Two-Photon Core >> Department of Neurosci& Physiology >> SUNY Upstate Medical University >> 766 Irving Ave., WH 3167 >> Syracuse, NY 13210 >> tel.: 315-464-7997 >> fax: 315-464-8014 >> email: [hidden email] >>>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear All, >> I wonder if someone could throw some light on a problem we have been > having >> for the past few years! We use a Leica SP5 confocal on an upright > microscope >> with dipping lenses to image pH and calcium-sensitive fluorescent dyes. > During >> solution changes - most notably the addition and removal of organic >> compounds such as propionate (20 mM)- we get severe image distortion and >> changes in focal depth. I attempted to attach a graph of a ROI from a fixed >> sealed specimen which we have imaged during solution changes with 3 >> different objectives - however, the ListServer rejected every image format I >> tried...I can send it by email if anyone is interested. It is impossible that the >> propionate is actually getting to the specimen itself - I suspect it must be as > a >> result of refractive index changes. >> >> My questions are: How do others image small structures (i.e. dendritic spines >> or NMJs) during changes of solution with differing refractive indicies? Is there >> some well know trick? Or, am I using the wrong objectives? Does wide field >> microscopy have the same problems? >> Many thanks, >> Christof Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology& Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
![]() ![]() |
Glen MacDonald-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I second Mark's statement. Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Cellular Morphology Core Center on Human Development and Disability Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] [hidden email] On Mar 19, 2013, at 8:08 AM, Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > It looks like a movement problem to me… > > Cheers > > On 19/03/2013, at 3:02 PM, Christof Schwiening <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear Arvydas, >> The graph is here: https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg >> I have tried three long working distance objectives (20-63X). I don't have a >> good example of the distorted image since it shifts faster than my resonnant >> scanner can capture it. However, the ROI plot shows the magnitude of the >> shifts on the fluorescence signal. We do see the same effect on fixed samples >> (which is what the image above shows). I am sure that it is some kind of >> Schlieren type of effect (refractive index). I think that in areas of high >> contrast even small optical deviations cause quite marked intensity changes..... >> Greetings, >> Christof >> >> On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas >> <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi Christof, >>> >>> More quantitative description of your imaging geometry/optics >>> would be useful. I have been observing similar effects when imaging >>> a thick sample on >>> inverted scope using 40x water dipping lens in immersion mode. >>> When water drop substantially shrinks after 2-3 hrs >>> focus starts shifting until disappears completely (from the set range). >>> This corresponds to ultimate refractive index change from 1.333 to 1.0. >>> The possible solution is to set big enough thickness of z-stack so that >>> your specimen still fits into imaged focal range (taking into account the >>> z-range shift due to refractive index change). >>> >>> I would like to see your graph and the distorted image. I wonder >>> if your chemicals are not dissolving well or recrystalizing thus creating >>> nonhomogenous layers distorting your image. Or maybe some vibrations >>> create inhomogeneities in the immersion/dipping medium. Do you observe this >>> type of distortion with fixed specimens as well? >>> >>> Best regards, >>> Arvydas >>> -------------------------------- >>> >>> >>> >>> >>> Arvydas Matiukas, Ph.D. >>> Director of Confocal&Two-Photon Core >>> Department of Neurosci& Physiology >>> SUNY Upstate Medical University >>> 766 Irving Ave., WH 3167 >>> Syracuse, NY 13210 >>> tel.: 315-464-7997 >>> fax: 315-464-8014 >>> email: [hidden email] >>>>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear All, >>> I wonder if someone could throw some light on a problem we have been >> having >>> for the past few years! We use a Leica SP5 confocal on an upright >> microscope >>> with dipping lenses to image pH and calcium-sensitive fluorescent dyes. >> During >>> solution changes - most notably the addition and removal of organic >>> compounds such as propionate (20 mM)- we get severe image distortion and >>> changes in focal depth. I attempted to attach a graph of a ROI from a fixed >>> sealed specimen which we have imaged during solution changes with 3 >>> different objectives - however, the ListServer rejected every image format I >>> tried...I can send it by email if anyone is interested. It is impossible that the >>> propionate is actually getting to the specimen itself - I suspect it must be as >> a >>> result of refractive index changes. >>> >>> My questions are: How do others image small structures (i.e. dendritic spines >>> or NMJs) during changes of solution with differing refractive indicies? Is there >>> some well know trick? Or, am I using the wrong objectives? Does wide field >>> microscopy have the same problems? >>> Many thanks, >>> Christof > > Mark B. Cannell Ph.D. FRSNZ > Professor of Cardiac Cell Biology > School of Physiology& Pharmacology > Medical Sciences Building > University of Bristol > Bristol > BS8 1TD UK > > [hidden email] |
![]() ![]() |
Arvydas Matiukas |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Christof , whatever is the reason for image distortion (Schlieren or mechanical movement) why don't you exclude the changing solution from the optical path, i.e. image your cells from below while adding solution from the top. Any confocal on an inverted stand would allow you to do this (you may need to switch to a dish with a coverslip window at the center). If switching to the inverted scope is not feasable/practical then try to improve your injection/perfusion. I recall my cell perfusion experiments on an inverted scope (some with simultaneous patch clamp), and the perfusion practically never was a problem both for image or patch clamp under condition that perfusion is properly designed ( flow <1ml/min, opening diameter ~0.5 mm and located few mm apart from the imaged/clamped cells. observing). Moreover, some organic compounds do not dissolve well (eg. when dissolving BA or BB in ethanol you can see Schlieren effect by a naked eye for several minutes). Try to very well dissolve (vortex?) your compound at elevated temperature, or try more gradual (slower) addition and replacement of the compound. Regarding possible formation of bubbles try cool the solution below room temperature ( at elevated temperatures gas solubility in liquid decreases and bubbles are not produced. Maybe your compound releases a lot of heat during dissolution, or changes the surfaces tension - if you introduce it more gradually all spatial inhomogeneties should be reduced. My first choice in your situation would be to try an inverted configuration. Best, Arvydas >>> Christof Schwiening <[hidden email]> 3/19/2013 10:02 AM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Arvydas, The graph is here: https://dl.dropbox.com/u/24087924/Convallaria%20slide.jpg I have tried three long working distance objectives (20-63X). I don't have a good example of the distorted image since it shifts faster than my resonnant scanner can capture it. However, the ROI plot shows the magnitude of the shifts on the fluorescence signal. We do see the same effect on fixed samples (which is what the image above shows). I am sure that it is some kind of Schlieren type of effect (refractive index). I think that in areas of high contrast even small optical deviations cause quite marked intensity changes..... Greetings, Christof On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi Christof, > >More quantitative description of your imaging geometry/optics >would be useful. I have been observing similar effects when imaging >a thick sample on >inverted scope using 40x water dipping lens in immersion mode. >When water drop substantially shrinks after 2-3 hrs >focus starts shifting until disappears completely (from the set range). >This corresponds to ultimate refractive index change from 1.333 to 1.0. >The possible solution is to set big enough thickness of z-stack so that >your specimen still fits into imaged focal range (taking into account the >z-range shift due to refractive index change). > >I would like to see your graph and the distorted image. I wonder >if your chemicals are not dissolving well or recrystalizing thus creating >nonhomogenous layers distorting your image. Or maybe some vibrations >create inhomogeneities in the immersion/dipping medium. Do you observe this >type of distortion with fixed specimens as well? > >Best regards, >Arvydas >-------------------------------- > > > > >Arvydas Matiukas, Ph.D. >Director of Confocal&Two-Photon Core >Department of Neurosci& Physiology >SUNY Upstate Medical University >766 Irving Ave., WH 3167 >Syracuse, NY 13210 >tel.: 315-464-7997 >fax: 315-464-8014 >email: [hidden email] >>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear All, >I wonder if someone could throw some light on a problem we have been >for the past few years! We use a Leica SP5 confocal on an upright microscope >with dipping lenses to image pH and calcium-sensitive fluorescent dyes. During >solution changes - most notably the addition and removal of organic >compounds such as propionate (20 mM)- we get severe image distortion and >changes in focal depth. I attempted to attach a graph of a ROI from a fixed >sealed specimen which we have imaged during solution changes with 3 >different objectives - however, the ListServer rejected every image format I >tried...I can send it by email if anyone is interested. It is impossible that the >propionate is actually getting to the specimen itself - I suspect it must be as a >result of refractive index changes. > >My questions are: How do others image small structures (i.e. dendritic spines >or NMJs) during changes of solution with differing refractive indicies? Is there >some well know trick? Or, am I using the wrong objectives? Does wide field >microscopy have the same problems? >Many thanks, >Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I really do doubt that it is real physical movement, but I will test it. If it is movement then there are only two things that could be moving: the objective or the glass slide. There is not much I can do about the objective but I can take greater steps to make sure that the glass slide does not move. It is already clamped into a standard microscope X-Y stage - but the glass slide might flex or the whole stage might move. I will attempt to test this by directly attaching the slide to the microscope. Although why this might only happen when 20 mM propionate (or other organic) is washed into, or out of the bath I have no idea. The movement is not coincident with the valves switching solutions but occurs with the appropriate delay for the arrival of the new solution from the manifold. There are no air bubbles and the movement does not occur when I switch between solution lines containing the same solution.... |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Arvydas, I agree that the problem would not be visible on an invert because the light does not travel through the mixing solution. Unfortunately we are imaging both NMJs in vivo (Drosophila larvae) and brain slices - both require good optics (both for patching and for imaging) on the top side of a relatively thick preparation. I can see that speeding up the solution flow would restrict the time the artefacts occur for, but we are already flowing at over 1 ml/min with a bath volume that is no greater than the bubble of solution that spans the preparation and objective (i.e. minimum volume, probably less than 100 uL). I think the problem is that we are not adequately mixing the whole solution space between the objective and preparation such that even with good flow rates a full solution change takes ~30s (about 5 bath volumes?). We want to image the pH changes during the instantaneous addition of weak acids and bases. The best we have managed so far is to add the solution directly to the site of interest using a puffer pipette. Greetings, Christof On Tue, 19 Mar 2013 14:33:08 -0400, Arvydas Matiukas <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi Christof , > >whatever is the reason for image distortion (Schlieren or mechanical movement) >why don't you exclude the changing solution from the optical path, i.e. image your >cells from below while adding solution from the top. Any confocal on an inverted >stand would allow you to do this (you may need to switch to a dish with a coverslip >window at the center). > >If switching to the inverted scope is not feasable/practical then try to improve your >injection/perfusion. I recall my cell perfusion experiments on an inverted scope > (some with simultaneous patch clamp), and the perfusion practically never was >a problem both for image or patch clamp under condition that perfusion is properly >designed ( flow <1ml/min, opening diameter ~0.5 mm and located few mm apart >from the imaged/clamped cells. observing). > >Moreover, some organic compounds do not dissolve well (eg. when dissolving BA >or BB in ethanol you can see Schlieren effect by a naked eye for several minutes). >Try to very well dissolve (vortex?) your compound at elevated temperature, or try >more gradual (slower) addition and replacement of the compound. Regarding possible >formation of bubbles try cool the solution below room temperature ( at elevated temperatures >gas solubility in liquid decreases and bubbles are not produced. Maybe your compound >releases a lot of heat during dissolution, or changes the surfaces tension - if you introduce >it more gradually all spatial inhomogeneties should be reduced. > >My first choice in your situation would be to try an inverted configuration. > >Best, >Arvydas > > >>>> Christof Schwiening <[hidden email]> 3/19/2013 10:02 AM >>> >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear Arvydas, >The graph is here: <a href="https://dl.dropbox.com/u/24087924/Convallaria%">https://dl.dropbox.com/u/24087924/Convallaria% >I have tried three long working distance objectives (20-63X). I don't have a >good example of the distorted image since it shifts faster than my resonnant >scanner can capture it. However, the ROI plot shows the magnitude of the >shifts on the fluorescence signal. We do see the same effect on fixed samples >(which is what the image above shows). I am sure that it is some kind of >Schlieren type of effect (refractive index). I think that in areas of high >contrast even small optical deviations cause quite marked intensity changes..... >Greetings, >Christof > >On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas ><[hidden email]> wrote: > >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>***** >> >>Hi Christof, >> >>More quantitative description of your imaging geometry/optics >>would be useful. I have been observing similar effects when imaging >>a thick sample on >>inverted scope using 40x water dipping lens in immersion mode. >>When water drop substantially shrinks after 2-3 hrs >>focus starts shifting until disappears completely (from the set range). >>This corresponds to ultimate refractive index change from 1.333 to 1.0. >>The possible solution is to set big enough thickness of z-stack so that >>your specimen still fits into imaged focal range (taking into account the >>z-range shift due to refractive index change). >> >>I would like to see your graph and the distorted image. I wonder >>if your chemicals are not dissolving well or recrystalizing thus creating >>nonhomogenous layers distorting your image. Or maybe some vibrations >>create inhomogeneities in the immersion/dipping medium. Do you observe this >>type of distortion with fixed specimens as well? >> >>Best regards, >>Arvydas >>-------------------------------- >> >> >> >> >>Arvydas Matiukas, Ph.D. >>Director of Confocal&Two-Photon Core >>Department of Neurosci& Physiology >>SUNY Upstate Medical University >>766 Irving Ave., WH 3167 >>Syracuse, NY 13210 >>tel.: 315-464-7997 >>fax: 315-464-8014 >>email: [hidden email] >>>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >>***** >>To join, leave or search the confocal microscopy listserv, go to: >>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>***** >> >>Dear All, >>I wonder if someone could throw some light on a problem we have been >having >>for the past few years! We use a Leica SP5 confocal on an upright >microscope >>with dipping lenses to image pH and calcium-sensitive fluorescent dyes. >During >>solution changes - most notably the addition and removal of organic >>compounds such as propionate (20 mM)- we get severe image distortion >>changes in focal depth. I attempted to attach a graph of a ROI from a fixed >>sealed specimen which we have imaged during solution changes with 3 >>different objectives - however, the ListServer rejected every image format I >>tried...I can send it by email if anyone is interested. It is impossible that the >>propionate is actually getting to the specimen itself - I suspect it must be as >a >>result of refractive index changes. >> >>My questions are: How do others image small structures (i.e. dendritic spines >>or NMJs) during changes of solution with differing refractive indicies? Is there >>some well know trick? Or, am I using the wrong objectives? Does wide field >>microscopy have the same problems? >>Many thanks, >>Christof |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Mark, I attached the glass slide to the base of my stage and tried the solution change again. This time I attempted to setup the optics so that I can video the effects (DIC). I have uploaded a video here: https://dl.dropbox.com/u/24087924/propionate.wmv A few seconds into the video I switch the perfusion valve to add the 20 mM propionate - it does cause a small instantaneous jump which is only just visible. The propionate then washes in and the effect can most clearly be seen from 23s after which I repeatedly stop and start the solution flow. I think the phenomenon that is visible under DIC must be the same one causing the confocal distortions. Does it look like you would expect for physical movement? Greetings, Christof On Tue, 19 Mar 2013 15:08:25 +0000, Mark Cannell <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >It looks like a movement problem to me… > >Cheers > >On 19/03/2013, at 3:02 PM, Christof Schwiening <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear Arvydas, >> The graph is here: <a href="https://dl.dropbox.com/u/24087924/Convallaria%">https://dl.dropbox.com/u/24087924/Convallaria% >> I have tried three long working distance objectives (20-63X). I don't have a >> good example of the distorted image since it shifts faster than my resonnant >> scanner can capture it. However, the ROI plot shows the magnitude of the >> shifts on the fluorescence signal. We do see the same effect on fixed samples >> (which is what the image above shows). I am sure that it is some kind of >> Schlieren type of effect (refractive index). I think that in areas of high >> contrast even small optical deviations cause quite marked intensity changes..... >> Greetings, >> Christof >> >> On Mon, 18 Mar 2013 16:25:51 -0400, Arvydas Matiukas >> <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi Christof, >>> >>> More quantitative description of your imaging geometry/optics >>> would be useful. I have been observing similar effects when imaging >>> a thick sample on >>> inverted scope using 40x water dipping lens in immersion mode. >>> When water drop substantially shrinks after 2-3 hrs >>> focus starts shifting until disappears completely (from the set range). >>> This corresponds to ultimate refractive index change from 1.333 to 1.0. >>> The possible solution is to set big enough thickness of z-stack so that >>> your specimen still fits into imaged focal range (taking into account the >>> z-range shift due to refractive index change). >>> >>> I would like to see your graph and the distorted image. I wonder >>> if your chemicals are not dissolving well or recrystalizing thus creating >>> nonhomogenous layers distorting your image. Or maybe some vibrations >>> create inhomogeneities in the immersion/dipping medium. Do you observe >>> type of distortion with fixed specimens as well? >>> >>> Best regards, >>> Arvydas >>> -------------------------------- >>> >>> >>> >>> >>> Arvydas Matiukas, Ph.D. >>> Director of Confocal&Two-Photon Core >>> Department of Neurosci& Physiology >>> SUNY Upstate Medical University >>> 766 Irving Ave., WH 3167 >>> Syracuse, NY 13210 >>> tel.: 315-464-7997 >>> fax: 315-464-8014 >>> email: [hidden email] >>>>>> Christof Schwiening <[hidden email]> 3/18/2013 10:29 AM >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear All, >>> I wonder if someone could throw some light on a problem we have been >> having >>> for the past few years! We use a Leica SP5 confocal on an upright >> microscope >>> with dipping lenses to image pH and calcium-sensitive fluorescent dyes. >> During >>> solution changes - most notably the addition and removal of organic >>> compounds such as propionate (20 mM)- we get severe image distortion >>> changes in focal depth. I attempted to attach a graph of a ROI from a fixed >>> sealed specimen which we have imaged during solution changes with 3 >>> different objectives - however, the ListServer rejected every image format I >>> tried...I can send it by email if anyone is interested. It is impossible that the >>> propionate is actually getting to the specimen itself - I suspect it must be as >> a >>> result of refractive index changes. >>> >>> My questions are: How do others image small structures (i.e. dendritic spines >>> or NMJs) during changes of solution with differing refractive indicies? Is there >>> some well know trick? Or, am I using the wrong objectives? Does wide field >>> microscopy have the same problems? >>> Many thanks, >>> Christof > >Mark B. Cannell Ph.D. FRSNZ >Professor of Cardiac Cell Biology >School of Physiology& Pharmacology >Medical Sciences Building >University of Bristol >Bristol >BS8 1TD UK > >[hidden email] |
![]() ![]() |
Glen MacDonald-2 |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Christof, That movie looked like motion and it seemed oscillatory. Is that a peristaltic pump with a stepper motor? Other questions would be whether there is a temperature difference between the propionate solution and the other solutions; size of tubing or differences in tubing adapters and couplers that may affect velocity or pressure, or differences in how the tubing is looped? what is the shape of your chamber? Diamond shaped chambers seem to have better mixing than round or square. Improved mixing may allow a reduction in the flow rate. Maybe test a reduced flow rate and see how that alters the shift, and compare to gravity flow. Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Cellular Morphology Core Center on Human Development and Disability Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] [hidden email] On Mar 20, 2013, at 4:17 AM, Christof Schwiening <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I really do doubt that it is real physical movement, but I will test it. If it is > movement then there are only two things that could be moving: the objective > or the glass slide. There is not much I can do about the objective but I can > take greater steps to make sure that the glass slide does not move. It is > already clamped into a standard microscope X-Y stage - but the glass slide > might flex or the whole stage might move. I will attempt to test this by > directly attaching the slide to the microscope. Although why this might only > happen when 20 mM propionate (or other organic) is washed into, or out of > the bath I have no idea. The movement is not coincident with the valves > switching solutions but occurs with the appropriate delay for the arrival of the > new solution from the manifold. There are no air bubbles and the movement > does not occur when I switch between solution lines containing the same > solution.... |
![]() ![]() |
Christof Schwiening |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Glen, The solutions are gravity-fed via independent tubes from the base of 50 ml syringes suspended about 50cm above the microscope stage. The tubes are identical and are connected to electrical on-off valves (General Valves). The tubes meet at a manifold about 15 cm from the output. In the case of the video the output is directly onto the top of a fixed-preparation (Convallaria) slide which was mounted directly onto the fixed platform of the microscope. The output tube was about 2 mm from the objective at the level of the slide. The suction was similarly sighted about 2 mm from the objective at an angle of 180 degrees to the input. The suction was from a bottle (2 L) maintained a low pressure with a high frequency low volume pump. The only oscillations that occur within the system are surface tension effects on the suction when air mixes with the water in the output lines - this occurs all of the time and sounds like a child sucking through a straw from an empty glass of lemonade! I doubt that it is movement. I guess I would need a force transducer mounted on the slide to prove that. Greetings, Christof |
![]() ![]() |
Mark Cannell-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You could use reflection mode from the surface of the slide to see if it's moving? Cheers On 21/03/2013, at 10:40 AM, Christof Schwiening <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Glen, > The solutions are gravity-fed via independent tubes from the base of 50 ml > syringes suspended about 50cm above the microscope stage. The tubes are > identical and are connected to electrical on-off valves (General Valves). The > tubes meet at a manifold about 15 cm from the output. In the case of the > video the output is directly onto the top of a fixed-preparation (Convallaria) > slide which was mounted directly onto the fixed platform of the microscope. > The output tube was about 2 mm from the objective at the level of the slide. > The suction was similarly sighted about 2 mm from the objective at an angle of > 180 degrees to the input. The suction was from a bottle (2 L) maintained a > low pressure with a high frequency low volume pump. The only oscillations that > occur within the system are surface tension effects on the suction when air > mixes with the water in the output lines - this occurs all of the time and > sounds like a child sucking through a straw from an empty glass of lemonade! > I doubt that it is movement. I guess I would need a force transducer mounted > on the slide to prove that. > Greetings, > Christof Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology& Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
![]() ![]() |
Stanislav Vitha-2 |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** After looking at the video, the "movements" look to me like a Schlieren effect; A wave of higher-RI solution rolls over the sample from left to right. Do you know what is the difference (if any) in refractive indices between the two solutions that you are switching? If that is indeed the effect of RI, you should not see any movements when you use two identical solutions and perform the same valve switching. If the movement is still there, A change in RI will cause Z-axis compression/expansion (focus shift); Perhaps you could tune the first buffer to the same RI as the propionate solution by some "inert" additive. BSA, gelatine, or some other non-osmotic agent comes to mind (but I know nothing about what might interfere with your samples). I am just guessing here, but another explanation may be that switching to the propionate buffer ('open-system buffer') enables osmotic effects of other media components and leads to cell shrinkage and intracellular pH changes. Something like in this paper: Thomas A. Heming, Gregory Boyarsky, Divina M. Tuazon, and Akhil Bidani: pHi responses to osmotic cell shrinkage in the presence of open-system buffers J Appl Physiol October 1, 2000 89:(4) 1543- 1552 Stan Vitha Microscopy and Imaging Center Texas A&M University |
![]() ![]() |
Glen MacDonald-2 |
![]() |
In reply to this post by Christof Schwiening
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** From your description, it seems to be a well designed flow system. But, the motion seemed to be oscillatory, which made me think it had a hydraulic cause. If Schlieren effects, an eddy in the chamber may be causing the higher RI solution to move back and forth as it mixes. Switching between similar solutions, as proposed by Stan, (including 2 propionates) would be a good test for motion. Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Cellular Morphology Core Center on Human Development and Disability Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] [hidden email] On Mar 21, 2013, at 3:40 AM, Christof Schwiening <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Glen, > The solutions are gravity-fed via independent tubes from the base of 50 ml > syringes suspended about 50cm above the microscope stage. The tubes are > identical and are connected to electrical on-off valves (General Valves). The > tubes meet at a manifold about 15 cm from the output. In the case of the > video the output is directly onto the top of a fixed-preparation (Convallaria) > slide which was mounted directly onto the fixed platform of the microscope. > The output tube was about 2 mm from the objective at the level of the slide. > The suction was similarly sighted about 2 mm from the objective at an angle of > 180 degrees to the input. The suction was from a bottle (2 L) maintained a > low pressure with a high frequency low volume pump. The only oscillations that > occur within the system are surface tension effects on the suction when air > mixes with the water in the output lines - this occurs all of the time and > sounds like a child sucking through a straw from an empty glass of lemonade! > I doubt that it is movement. I guess I would need a force transducer mounted > on the slide to prove that. > Greetings, > Christof |