To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, Got a weird issue I cannot explain. We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed slides, on Leica widefield with scanner. We are getting great initial bleaching of the YFP acceptor, however, there is an almost immediate recovery of fluorescence in the bleached ROI. This is happening also in the YFP only control. Can this possibly be because of the mounting medium? What else can be a mobile fraction on a fixed slide? The signal is good, has the expected expression pattern. Cheers, Avi -- Avi Jacob, Ph.D. Head of The Kanbar Light Microscopy Unit Bar-Ilan University, Ramat-Gan 5290002, Israel http://tinyurl.com/BIU-Microscopy |
![]() ![]() |
Steffen Dietzel |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Avi, that is weird. I would try to figure out if this is due to diffusion of unbleached YFP (or some contaminant in the mounting medium) or due to actual recovery of dark proteins becoming fluorescent again. Bleaching of a large area might give a clue, or a FLIP experiment. Good hunting Steffen Am 19.07.2018 um 09:33 schrieb Avi Jacob: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > Got a weird issue I cannot explain. > > We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed > slides, on Leica widefield with scanner. > We are getting great initial bleaching of the YFP acceptor, however, there > is an almost immediate recovery of fluorescence in the bleached ROI. > This is happening also in the YFP only control. > > Can this possibly be because of the mounting medium? What else can be a > mobile fraction on a fixed slide? > The signal is good, has the expected expression pattern. > > Cheers, > Avi > > -- > Avi Jacob, Ph.D. > Head of The Kanbar Light Microscopy Unit > Bar-Ilan University, Ramat-Gan 5290002, Israel > http://tinyurl.com/BIU-Microscopy > ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Biomedical Center (BMC) Head of the Core Facility Bioimaging Großhaderner Straße 9 D-82152 Planegg-Martinsried Germany http://www.bioimaging.bmc.med.uni-muenchen.de |
![]() ![]() |
Kevin Dean |
![]() |
In reply to this post by Avi Jacob
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** This is a common problem with fluorescent proteins. Reversible photobleaching, or dark-state conversion. You can find quite a few papers on the topic once you know the keywords to search for. One shameless self-reference: https://www.sciencedirect.com/science/article/pii/S0006349511007867 If you are working with fixed specimens, I would recommend just photobleaching for longer. Best, Kevin On Jul 19, 2018, at 2:33 AM, Avi Jacob <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, Got a weird issue I cannot explain. We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed slides, on Leica widefield with scanner. We are getting great initial bleaching of the YFP acceptor, however, there is an almost immediate recovery of fluorescence in the bleached ROI. This is happening also in the YFP only control. Can this possibly be because of the mounting medium? What else can be a mobile fraction on a fixed slide? The signal is good, has the expected expression pattern. Cheers, Avi -- Avi Jacob, Ph.D. Head of The Kanbar Light Microscopy Unit Bar-Ilan University, Ramat-Gan 5290002, Israel http://tinyurl.com/BIU-Microscopy ________________________________ UT Southwestern Medical Center The future of medicine, today. |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by Avi Jacob
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Avi, FRET of FP pairs after fixation is a complete waste of time (don't believe any literature that claims it is useful -- even if authored by confocal listserv members). By extension, FRET between fluorescent antibodies (Fab's or nanobodies would be smaller, so potentially bigger dynamic range) of epitopes detected after fixation may be subject to the same issue(s). The simple test is compare your live vs fixed by acceptor photobleaching. The best way to do FRET quantitation is FLIM on live cells (and again, waste of time to do FLIM-FRET measurements of FP's on fixed cells). Traditionally this has been donor FLIM only -- I suggest measuring donor and acceptor FLIM simultaneously is better, and several companies now offer 30+ million counts per second photon counting (see p.s.). George p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS FastFLIM (October?) demonstrations/workshops here at JHU. I'm also now (as of this email) thinking I should attend the "Advanced TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that Wolfgang Becker recently posted about. https://www.eventbrite.com/e/12th-annual-workshop-on-advanced-tcspc-techniques-in-the-biomedical-sciences-tickets-46820517428 TCSPC has historically been "painfully slow, small field of view" FLIM ( under 1 million photon counts per second ... which would be average ~1 photon per pixel per second if imaging 1000x1000 pixels, and more usefully ~100 counts per pixel per second for 100x100 pixels)). I am intrigued by FALCON, FastFLIM, and similar from Becker&Hickl and probably others (PicoQuant) because a lot faster (>=30 million counts per counting channel - different companies may have different ways to up the count rate). FYI - I am also intrigued by the potential of using "fast counting" electronics for simple photon counting at ~30x faster than I can do now on our current SP8 HyD's http://confocal.jhu.edu/current-equipment/leica-sp8-confocal-microscope (I have not updated FALCON visit dates - yet). That is, not -- yet -- investing in pulsed lasers and simply 'count faster'. On 7/19/2018 3:33 AM, Avi Jacob wrote: > ***** > confocal microscopy listserv > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > Got a weird issue I cannot explain. > > We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed > slides, on Leica widefield with scanner. > We are getting great initial bleaching of the YFP acceptor, however, there > is an almost immediate recovery of fluorescence in the bleached ROI. > This is happening also in the YFP only control. > > Can this possibly be because of the mounting medium? What else can be a > mobile fraction on a fixed slide? > The signal is good, has the expected expression pattern. > > Cheers, > Avi > > -- > Avi Jacob, Ph.D. > Head of The Kanbar Light Microscopy Unit > Bar-Ilan University, Ramat-Gan 5290002, Israel > http://tinyurl.com/BIU-Microscopy -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
![]() ![]() |
Martin Wessendorf-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hey, George! It'd be helpful if you could offer some references or give your reasoning for thinking that fixed-tissue FRET is a waste of time. Thanks! Martin Wessendorf On 7/19/2018 8:52 AM, George McNamara wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Avi, > > FRET of FP pairs after fixation is a complete waste of time (don't > believe any literature that claims it is useful -- even if authored by > confocal listserv members). By extension, FRET between fluorescent > antibodies (Fab's or nanobodies would be smaller, so potentially > bigger dynamic range) of epitopes detected after fixation may be > subject to the same issue(s). > > The simple test is compare your live vs fixed by acceptor photobleaching. > > The best way to do FRET quantitation is FLIM on live cells (and again, > waste of time to do FLIM-FRET measurements of FP's on fixed cells). > Traditionally this has been donor FLIM only -- I suggest measuring > donor and acceptor FLIM simultaneously is better, and several > companies now offer 30+ million counts per second photon counting (see > p.s.). > > > George > > p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS FastFLIM > (October?) demonstrations/workshops here at JHU. > > I'm also now (as of this email) thinking I should attend the "Advanced > TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that Wolfgang > Becker recently posted about. > > https://www.eventbrite.com/e/12th-annual-workshop-on-advanced-tcspc-techniques-in-the-biomedical-sciences-tickets-46820517428 > > > TCSPC has historically been "painfully slow, small field of view" FLIM > ( under 1 million photon counts per second ... which would be average > ~1 photon per pixel per second if imaging 1000x1000 pixels, and more > usefully ~100 counts per pixel per second for 100x100 pixels)). I am > intrigued by FALCON, FastFLIM, and similar from Becker&Hickl and > probably others (PicoQuant) because a lot faster (>=30 million counts > per counting channel - different companies may have different ways to > up the count rate). > > FYI - I am also intrigued by the potential of using "fast counting" > electronics for simple photon counting at ~30x faster than I can do > now on our current SP8 HyD's > http://confocal.jhu.edu/current-equipment/leica-sp8-confocal-microscope > (I have not updated FALCON visit dates - yet). That is, not -- yet -- > investing in pulsed lasers and simply 'count faster'. > > > On 7/19/2018 3:33 AM, Avi Jacob wrote: >> ***** >> confocal microscopy listserv >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> Hi all, >> >> Got a weird issue I cannot explain. >> >> We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed >> slides, on Leica widefield with scanner. >> We are getting great initial bleaching of the YFP acceptor, however, >> there >> is an almost immediate recovery of fluorescence in the bleached ROI. >> This is happening also in the YFP only control. >> >> Can this possibly be because of the mounting medium? What else can be a >> mobile fraction on a fixed slide? >> The signal is good, has the expected expression pattern. >> >> Cheers, >> Avi >> >> -- >> Avi Jacob, Ph.D. >> Head of The Kanbar Light Microscopy Unit >> Bar-Ilan University, Ramat-Gan 5290002, Israel >> http://tinyurl.com/BIU-Microscopy > -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
![]() ![]() |
Periasamy, Ammasi (ap3t)-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** George, I do not agree with you, it's waste of time to do FRET with fixed cells. As you know there are number of papers published on acceptor photobleaching not only in live cells but also in fixed cells. For example, RNA-protein interaction there is no way you can work with live cells. Rehman, S., Gladman, J.T., Periasamy, A., Sun, Y. and Mahadevan, M.S. (2014) Development of an AP-FRET based analysis for characterinzing RNA-protein interactions in myotonic dystrophy (DM1). PLoS ONE 9(4):e95957. On FLIM-FRET, we have tested for fixed cells (not published), the lifetime changes only after a week or two of fixation. Yes, FLIM has to be used for live specimens. For some experiment you do not have a choice except to use fixed cells for FRET. Hope this helps. Ammasi Dr. Ammasi Periasamy Professor & Center Director, WM Keck Center for Cellular Imaging, Department of Biology, Univeristy of Virginia, 90 Geldard Dr., Charlottesville, VA 22904, USA. http://www.kcci.virginia.edu/people/profile/ap3t Phone: (434) 243-7602 or 982-4869 Fax: (434) 982-5210 E-mail: [hidden email] FRET/FLIM Workshop-March 11-15, 2019: http://www.kcci.virginia.edu/workshop -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Martin Wessendorf Sent: Thursday, July 19, 2018 10:57 AM To: [hidden email] Subject: Re: Weird issue with FRET after acceptor PB ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hey, George! It'd be helpful if you could offer some references or give your reasoning for thinking that fixed-tissue FRET is a waste of time. Thanks! Martin Wessendorf On 7/19/2018 8:52 AM, George McNamara wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Avi, > > FRET of FP pairs after fixation is a complete waste of time (don't > believe any literature that claims it is useful -- even if authored by > confocal listserv members). By extension, FRET between fluorescent > antibodies (Fab's or nanobodies would be smaller, so potentially > bigger dynamic range) of epitopes detected after fixation may be > subject to the same issue(s). > > The simple test is compare your live vs fixed by acceptor photobleaching. > > The best way to do FRET quantitation is FLIM on live cells (and again, > waste of time to do FLIM-FRET measurements of FP's on fixed cells). > Traditionally this has been donor FLIM only -- I suggest measuring > donor and acceptor FLIM simultaneously is better, and several > companies now offer 30+ million counts per second photon counting (see > p.s.). > > > George > > p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS FastFLIM > (October?) demonstrations/workshops here at JHU. > > I'm also now (as of this email) thinking I should attend the "Advanced > TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that Wolfgang > Becker recently posted about. > > https://www.eventbrite.com/e/12th-annual-workshop-on-advanced-tcspc-te > chniques-in-the-biomedical-sciences-tickets-46820517428 > > > TCSPC has historically been "painfully slow, small field of view" FLIM > ( under 1 million photon counts per second ... which would be average > ~1 photon per pixel per second if imaging 1000x1000 pixels, and more > usefully ~100 counts per pixel per second for 100x100 pixels)). I am > intrigued by FALCON, FastFLIM, and similar from Becker&Hickl and > probably others (PicoQuant) because a lot faster (>=30 million counts > per counting channel - different companies may have different ways to > up the count rate). > > FYI - I am also intrigued by the potential of using "fast counting" > electronics for simple photon counting at ~30x faster than I can do > now on our current SP8 HyD's > http://confocal.jhu.edu/current-equipment/leica-sp8-confocal-microscop > e (I have not updated FALCON visit dates - yet). That is, not -- yet > -- investing in pulsed lasers and simply 'count faster'. > > > On 7/19/2018 3:33 AM, Avi Jacob wrote: >> ***** >> confocal microscopy listserv >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> Hi all, >> >> Got a weird issue I cannot explain. >> >> We are doing acceptor photobleaching FRET with a CFP-YFP pair, on >> fixed slides, on Leica widefield with scanner. >> We are getting great initial bleaching of the YFP acceptor, however, >> there is an almost immediate recovery of fluorescence in the bleached >> ROI. >> This is happening also in the YFP only control. >> >> Can this possibly be because of the mounting medium? What else can be >> a mobile fraction on a fixed slide? >> The signal is good, has the expected expression pattern. >> >> Cheers, >> Avi >> >> -- >> Avi Jacob, Ph.D. >> Head of The Kanbar Light Microscopy Unit Bar-Ilan University, >> Ramat-Gan 5290002, Israel http://tinyurl.com/BIU-Microscopy > -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by Martin Wessendorf-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Martin, thanks alot for the prompt. I am pasting below just selected -- by me -- text from the abstracts. I note that this (i) is not comprehensive, (ii) the texts are not in agreement (nor are the methods the same, re: fixation, mounting media, use or not of nail polish sealer). I'll also 'double down' up here on my "FRET of FP pairs after fixation is a complete waste of time" with: each cell should be used as its own control, in a time series, under identical conditions (live!) conditions, i.e. before or after stimulation. That is, every single cell, even if it is from an expanded clone, or even if a small clone, such as the 'HeLa tetrad' approach used in fig F2 of https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3012539/figure/F2/ (which was not a FRET paper, and I still like this approach to efficient quantitation of cell fluorescence. I am also going to point out up top here that for FP biosensors (the original post does not mention whether their experiments were FP biosensors, or simply proximity assay, or other) that single FP, i.e. GCaMP7 dark --> bright on Ca++ binding, is the future AND I am enthusiastic about the new dLight1 GPCR FP biosensor paper see https://www.ncbi.nlm.nih.gov/pubmed/29853555 (even though they used GCaMP6 instead of 7) because it implies every GPCR can be turned into a dLight-like FP biosensor (dLight1 optimation in main text, 7 un-optimized in supplemental) (note: I am not connected to the dLight work). OK, back to "fixed cell FRET is a waste of time" light survey: Reminder: I am selectively quoting from the abstracts. If anyone wants more, read the full abstract and papers. https://www.ncbi.nlm.nih.gov/pubmed/25179491 Effect of*fixation*procedures on the fluorescence lifetimes of Aequorea victoria derived fluorescent proteins. <https://www.ncbi.nlm.nih.gov/pubmed/25179491> Joosen L, Hink MA, Gadella TW Jr, Goedhart J. J Microsc. 2014 Dec;256(3):166-76. doi: 10.1111/jmi.12168. Epub 2014 Sep 1. PMID: 25179491 However, the effect offixation methods and mounting media on fluorescence lifetime is poorly documented so far. Here, we demonstrate thatfixationby formaldehyde or methanol itself does not affect the lifetime of fluorescent proteins produced in cells but that several widely used mounting media decrease the fluorescence lifetime by up to 20%. It is shown that fixed cells producing Aequorea victoria derived fluorescent proteins mounted in Tris buffer have fluorescence lifetimes indistinguishable from values measured in living cells. Tris buffer also allows accurate Förster Resonance Energy Transfer quantification in fixed cells, as shown with an mTurquoise2-SYFP2 fusion protein. Moreover, identical lifetime contrasts are measured in living and fixed cells mounted in Tris buffer after introducing a single plasmid expressing two lifetime variants of cyan fluorescent proteins, each targeted to different locations in the cell. https://www.ncbi.nlm.nih.gov/pubmed/?term=fret+fixation+artifacts (3 of 4 hits summarized below) https://www.ncbi.nlm.nih.gov/pubmed/21490932 Some secrets of fluorescent proteins: distinct bleaching in various mounting fluids and photoactivation of cyan fluorescent proteins at YFP-excitation. <https://www.ncbi.nlm.nih.gov/pubmed/21490932> Malkani N, Schmid JA. PLoS One. 2011 Apr 7;6(4):e18586. doi: 10.1371/journal.pone.0018586. PMID: 21490932 When we applied a commonly usedFRETmicroscopy technique--the increase in donor (CFP)-fluorescence after bleaching of acceptor fluorophores (YFP), we obtained good signals in live cells, but very weak signals for the same samples afterfixationand mounting in commercial microscopy mounting fluids. This observation could be traced back to much faster bleaching of CFP in these mounting media. Strikingly, the opposite effect of the mounting fluid was observed for YFP and also for other proteins such as Cerulean, TFP or Venus. The changes in photostability of CFP and YFP were not caused by thefixationbut directly dependent on the mounting fluid. Furthermore we made the interesting observation that the CFP-fluorescence intensity increases by about 10-15% after illumination at the YFP-excitation wavelength--a phenomenon, which was also observed for Cerulean. This photoactivation of cyan fluorescent proteins at the YFP-excitation can cause false-positive signals in theFRET-microscopy technique that is based on bleaching of a yellowFRETacceptor. https://www.ncbi.nlm.nih.gov/pubmed/18359791 Resonance energy transfer in cells: a new look at*fixation*effect and receptor aggregation on cell membrane. <https://www.ncbi.nlm.nih.gov/pubmed/18359791> Anikovsky M, Dale L, Ferguson S, Petersen N. Biophys J. 2008 Aug;95(3):1349-59. doi: 10.1529/biophysj.107.124313. Epub 2008 Mar 21. PMID: 18359791 We have considered systems in whichFREToccurs as intramolecular and/or intermolecular process. The proposed dynamicFRETmodel shows that in the case of intermolecular process the degree of aggregation only slightly affects the energy transfer efficiency. ...This dual dependence often leads to ambiguity. In this article, we show howFRETefficiency can be significantly reduced even in highly coupled system through conformational restrictions in the donor-acceptor pair. Importantly, such restrictions can be imposed on the system by cellfixation, a procedure routinely used when conductingFRETmeasurements. https://www.ncbi.nlm.nih.gov/pubmed/18453757 *Fixation*, mounting and sealing with nail polish of cell specimens lead to incorrect*FRET*measurements using acceptor photobleaching. <https://www.ncbi.nlm.nih.gov/pubmed/18453757> Rodighiero S, Bazzini C, Ritter M, Fürst J, Botta G, Meyer G, Paulmichl M. Cell Physiol Biochem. 2008;21(5-6):489-98. doi: 10.1159/000129642. Epub 2008 Apr 24. PMID: 18453757 ... the calculation of theFRETefficiency by de-quenching the donor cyan-fluorescent-protein (CFP) emission due to acceptor-photobleaching leads to erroneous estimate of theFRETefficiency in fixed, mounted and sealed specimens. The acceptor photobleaching leads to a modification of the donor cyan-fluorescent-protein, which shows then a strong emission, thus mimicking functional interaction between CFP (donor) and yellow-fluorescent-protein (YFP; acceptor). Moreover, the procedure of acceptor photobleaching masks physiological (non random) interaction between molecules within the fixed, mounted and sealed cell. // another weakness of fixed cell FRET is that it is not readily 'translatable' to either mouse (or other model systems) in vivo imaging, and especially not to human clinical fluorescence in vivo. enjoy, George p.s. above is all fluorescent proteins. Funky things can happen to small molecules. As one example, Cy3 can readily undergo (at times a researcher may not be thinking about ... lifetime could be used to measure) "Protein-inducedfluorescenceenhancement(PIFE)" -- see https://www.ncbi.nlm.nih.gov/pubmed/26263254 for an entry point to PIFE (the abstract does not make it clear that "cis-Cy3" is dark, "trans-Cy3" is the fluorescent form ... stated in body of that text). p.p.s. not that I'm recommending it (since the whole point of my posts to is to avoid doing "fixation FRET"), the "best" approach might be to do on the microscope (carefully): simply aspirate the medium, add appropriate dilution fixative, fix [optimum or at least consistent time?], wash, stick in a simple buffer or optiimum (?) pH (Joosen et al liked Tris, but abstract did not disclose pH) and measure. Maybe: thoroughly aerate the imaging medium (in some consistent way: good luck) to maximize O2. A better use of scope time: image cells live. On 7/19/2018 10:56 AM, Martin Wessendorf wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hey, George! > > It'd be helpful if you could offer some references or give your > reasoning for thinking that fixed-tissue FRET is a waste of time. > > Thanks! > > Martin Wessendorf > > > > > On 7/19/2018 8:52 AM, George McNamara wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> Hi Avi, >> >> FRET of FP pairs after fixation is a complete waste of time (don't >> believe any literature that claims it is useful -- even if authored >> by confocal listserv members). By extension, FRET between fluorescent >> antibodies (Fab's or nanobodies would be smaller, so potentially >> bigger dynamic range) of epitopes detected after fixation may be >> subject to the same issue(s). >> >> The simple test is compare your live vs fixed by acceptor >> photobleaching. >> >> The best way to do FRET quantitation is FLIM on live cells (and >> again, waste of time to do FLIM-FRET measurements of FP's on fixed >> cells). Traditionally this has been donor FLIM only -- I suggest >> measuring donor and acceptor FLIM simultaneously is better, and >> several companies now offer 30+ million counts per second photon >> counting (see p.s.). >> >> >> George >> >> p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS >> FastFLIM (October?) demonstrations/workshops here at JHU. >> >> I'm also now (as of this email) thinking I should attend the >> "Advanced TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that >> Wolfgang Becker recently posted about. >> >> https://www.eventbrite.com/e/12th-annual-workshop-on-advanced-tcspc-techniques-in-the-biomedical-sciences-tickets-46820517428 >> >> >> TCSPC has historically been "painfully slow, small field of view" >> FLIM ( under 1 million photon counts per second ... which would be >> average ~1 photon per pixel per second if imaging 1000x1000 pixels, >> and more usefully ~100 counts per pixel per second for 100x100 >> pixels)). I am intrigued by FALCON, FastFLIM, and similar from >> Becker&Hickl and probably others (PicoQuant) because a lot faster >> (>=30 million counts per counting channel - different companies may >> have different ways to up the count rate). >> >> FYI - I am also intrigued by the potential of using "fast counting" >> electronics for simple photon counting at ~30x faster than I can do >> now on our current SP8 HyD's >> http://confocal.jhu.edu/current-equipment/leica-sp8-confocal-microscope >> (I have not updated FALCON visit dates - yet). That is, not -- yet -- >> investing in pulsed lasers and simply 'count faster'. >> >> >> On 7/19/2018 3:33 AM, Avi Jacob wrote: >>> ***** >>> confocal microscopy listserv >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >>> posting. >>> ***** >>> >>> Hi all, >>> >>> Got a weird issue I cannot explain. >>> >>> We are doing acceptor photobleaching FRET with a CFP-YFP pair, on fixed >>> slides, on Leica widefield with scanner. >>> We are getting great initial bleaching of the YFP acceptor, however, >>> there >>> is an almost immediate recovery of fluorescence in the bleached ROI. >>> This is happening also in the YFP only control. >>> >>> Can this possibly be because of the mounting medium? What else can be a >>> mobile fraction on a fixed slide? >>> The signal is good, has the expected expression pattern. >>> >>> Cheers, >>> Avi >>> >>> -- >>> Avi Jacob, Ph.D. >>> Head of The Kanbar Light Microscopy Unit >>> Bar-Ilan University, Ramat-Gan 5290002, Israel >>> http://tinyurl.com/BIU-Microscopy >> > -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by Periasamy, Ammasi (ap3t)-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Peri, I am happy to discuss the merits of fixed cell FRET imaging at a bar/restaurant at M&M 2018, Aug 5-9, Baltimore, https://www.microscopy.org/mandm/2018 and, if I attend, the Bethesda Advanced TCSPC meeting in Bethesda in October. As for characterizing RNA-protein interactions, maybe FRET is not the optimal method. At least for genetically engineered systems (which practically all FP fusion protein experiments are), a different design may be better, for MS2/PP7 system(s), example: https://www.ncbi.nlm.nih.gov/pubmed/24402470 Background free imaging of single mRNAs in live cells using split fluorescent proteins. <https://www.ncbi.nlm.nih.gov/pubmed/24402470> Wu B, Chen J,*Singer*RH. Sci Rep. 2014 Jan 9;4:3615. doi: 10.1038/srep03615. PMID: 24402470 See also: https://www.ncbi.nlm.nih.gov/pubmed/29345990 Imaging mRNA In Vivo, from Birth to Death. <https://www.ncbi.nlm.nih.gov/pubmed/29345990> Tutucci E, Livingston NM,*Singer RH*,*Wu B*. Annu Rev Biophys. 2018 May 20;47:85-106. doi: 10.1146/annurev-biophys-070317-033037. Epub 2018 Jan 18. PMID: 29345990 https://www.ncbi.nlm.nih.gov/pubmed/27313041 Translation dynamics of single mRNAs in live cells and neurons. <https://www.ncbi.nlm.nih.gov/pubmed/27313041> *Wu B*, Eliscovich C, Yoon YJ,*Singer RH*. Science. 2016 Jun 17;352(6292):1430-5. doi: 10.1126/science.aaf1084. Epub 2016 May 5. PMID: 27313041 Disclosure: Prof. Bin Wu is now my supervisor at JHU (I am writing this on my own time). enjoy, George On 7/19/2018 11:43 AM, Periasamy, Ammasi (ap3t) wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > George, > I do not agree with you, it's waste of time to do FRET with fixed cells. As you know there are number of papers published on acceptor photobleaching not only in live cells but also in fixed cells. > For example, RNA-protein interaction there is no way you can work with live cells. > Rehman, S., Gladman, J.T., Periasamy, A., Sun, Y. and Mahadevan, M.S. (2014) Development of an AP-FRET based analysis for characterinzing RNA-protein interactions in myotonic dystrophy (DM1). PLoS ONE 9(4):e95957. > > On FLIM-FRET, we have tested for fixed cells (not published), the lifetime changes only after a week or two of fixation. Yes, FLIM has to be used for live specimens. For some experiment you do not have a choice except to use fixed cells for FRET. > Hope this helps. > Ammasi > > Dr. Ammasi Periasamy > Professor & Center Director, > WM Keck Center for Cellular Imaging, > Department of Biology, Univeristy of Virginia, > 90 Geldard Dr., Charlottesville, VA 22904, USA. > > http://www.kcci.virginia.edu/people/profile/ap3t > Phone: (434) 243-7602 or 982-4869 > Fax: (434) 982-5210 > E-mail: [hidden email] > > FRET/FLIM Workshop-March 11-15, 2019: http://www.kcci.virginia.edu/workshop > > > > -----Original Message----- > From: Confocal Microscopy List <[hidden email]> On Behalf Of Martin Wessendorf > Sent: Thursday, July 19, 2018 10:57 AM > To: [hidden email] > Subject: Re: Weird issue with FRET after acceptor PB > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hey, George! > > It'd be helpful if you could offer some references or give your reasoning for thinking that fixed-tissue FRET is a waste of time. > > Thanks! > > Martin Wessendorf > > > > > On 7/19/2018 8:52 AM, George McNamara wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi Avi, >> >> FRET of FP pairs after fixation is a complete waste of time (don't >> believe any literature that claims it is useful -- even if authored by >> confocal listserv members). By extension, FRET between fluorescent >> antibodies (Fab's or nanobodies would be smaller, so potentially >> bigger dynamic range) of epitopes detected after fixation may be >> subject to the same issue(s). >> >> The simple test is compare your live vs fixed by acceptor photobleaching. >> >> The best way to do FRET quantitation is FLIM on live cells (and again, >> waste of time to do FLIM-FRET measurements of FP's on fixed cells). >> Traditionally this has been donor FLIM only -- I suggest measuring >> donor and acceptor FLIM simultaneously is better, and several >> companies now offer 30+ million counts per second photon counting (see >> p.s.). >> >> >> George >> >> p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS FastFLIM >> (October?) demonstrations/workshops here at JHU. >> >> I'm also now (as of this email) thinking I should attend the "Advanced >> TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that Wolfgang >> Becker recently posted about. >> >> https://www.eventbrite.com/e/12th-annual-workshop-on-advanced-tcspc-te >> chniques-in-the-biomedical-sciences-tickets-46820517428 >> >> >> TCSPC has historically been "painfully slow, small field of view" FLIM >> ( under 1 million photon counts per second ... which would be average >> ~1 photon per pixel per second if imaging 1000x1000 pixels, and more >> usefully ~100 counts per pixel per second for 100x100 pixels)). I am >> intrigued by FALCON, FastFLIM, and similar from Becker&Hickl and >> probably others (PicoQuant) because a lot faster (>=30 million counts >> per counting channel - different companies may have different ways to >> up the count rate). >> >> FYI - I am also intrigued by the potential of using "fast counting" >> electronics for simple photon counting at ~30x faster than I can do >> now on our current SP8 HyD's >> http://confocal.jhu.edu/current-equipment/leica-sp8-confocal-microscop >> e (I have not updated FALCON visit dates - yet). That is, not -- yet >> -- investing in pulsed lasers and simply 'count faster'. >> >> >> On 7/19/2018 3:33 AM, Avi Jacob wrote: >>> ***** >>> confocal microscopy listserv >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >>> posting. >>> ***** >>> >>> Hi all, >>> >>> Got a weird issue I cannot explain. >>> >>> We are doing acceptor photobleaching FRET with a CFP-YFP pair, on >>> fixed slides, on Leica widefield with scanner. >>> We are getting great initial bleaching of the YFP acceptor, however, >>> there is an almost immediate recovery of fluorescence in the bleached >>> ROI. >>> This is happening also in the YFP only control. >>> >>> Can this possibly be because of the mounting medium? What else can be >>> a mobile fraction on a fixed slide? >>> The signal is good, has the expected expression pattern. >>> >>> Cheers, >>> Avi >>> >>> -- >>> Avi Jacob, Ph.D. >>> Head of The Kanbar Light Microscopy Unit Bar-Ilan University, >>> Ramat-Gan 5290002, Israel http://tinyurl.com/BIU-Microscopy -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
![]() ![]() |
Nicolai.Urban@mpfi.org |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Avi, getting back to your initial question (fluorescence recovery after photobleaching), there might be an explanation beyond having a mobile fraction of dye in your fixated sample. Have you considered that instead of actually (irreversibly) photobleaching the YFP, you are actually pumping it into a long-lived dark state, from which it can recover, given time? YFP has been shown to exhibit these dark-states (added some links below), which expresses itself as blinking or as recovery of fluorescence over time/after near-UV irradiation. It is fairly common for numerous FPs, and could explain the recovery despite everything being fixated and immobile. You could test it by either trying to induce recovery with a 405 nm laser (or similar), or possibly trying to bleach using much harsher illumination. Best of luck with your experiments,! Nicolai Urban >>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<< Dr. Nicolai T. Urban Max Planck Florida Institute Jupiter, 33458 FL, USA >>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<< Links concerning YFP dark-states: Ha & Tinnefeld; Photophysics of Fluorescence Probes for Single Molecule Biophysics and Super-Resolution Imaging; Annu Rev Phys Chem. 2012; 63: 595–617.; 10.1146/annurev-physchem-032210-103340 https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3736144/ Dickson RM, Cubitt AB, Tsien RY, Moerner WE. On/off blinking and switching behavior of single molecules of green fluorescent protein. Nature. 1997;388:355–358. Biteen JS, Thompson MA, Tselentis NK, Bowman GR, Shapiro L, Moerner WE. Super-resolution imaging in live Caulobacter crescentus cells using photoswitchable EYFP. Nat Methods. 2008;5:947–949. https://www.ncbi.nlm.nih.gov/pmc/articles/PMC2655310/ >>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<<>>>>>>>>>><<<<<<<<<< -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of George McNamara Sent: Donnerstag, 19. Juli 2018 12:32 To: [hidden email] Subject: Re: Weird issue with FRET after acceptor PB ***** To join, leave or search the confocal microscopy listserv, go to: https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293404512&sdata=NBx0ticPz9cCs68Tplgc6misH05lf%2Bqupn7oHAWZ%2Bok%3D&reserved=0 Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293404512&sdata=DXc%2B7aVGZu%2F3pSrpyW8ACo71J87My%2F9f0vEy7EThlWI%3D&reserved=0 and include the link in your posting. ***** Hi Peri, I am happy to discuss the merits of fixed cell FRET imaging at a bar/restaurant at M&M 2018, Aug 5-9, Baltimore, https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.microscopy.org%2Fmandm%2F2018&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293404512&sdata=cjJEItNF6ukYY%2BtvtX0hjEEi1OdW6DURH5U6XOW43rg%3D&reserved=0 and, if I attend, the Bethesda Advanced TCSPC meeting in Bethesda in October. As for characterizing RNA-protein interactions, maybe FRET is not the optimal method. At least for genetically engineered systems (which practically all FP fusion protein experiments are), a different design may be better, for MS2/PP7 system(s), example: https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F24402470&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293404512&sdata=%2BLMIBh1CGkrW3eTff31ls3H4c%2B768H%2BuwmmgVzOYcrU%3D&reserved=0 Background free imaging of single mRNAs in live cells using split fluorescent proteins. <https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F24402470&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=IxJNdL9WCblXiyJI5H%2FhyZ%2FGjf9GHeBOxohMEhJyimE%3D&reserved=0> Wu B, Chen J,*Singer*RH. Sci Rep. 2014 Jan 9;4:3615. doi: 10.1038/srep03615. PMID: 24402470 See also: https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F29345990&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=xXBHnrtX6eeRtyhETHL6%2BjjxZCdmYp7HEory%2FRVs4n8%3D&reserved=0 Imaging mRNA In Vivo, from Birth to Death. <https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F29345990&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=xXBHnrtX6eeRtyhETHL6%2BjjxZCdmYp7HEory%2FRVs4n8%3D&reserved=0> Tutucci E, Livingston NM,*Singer RH*,*Wu B*. Annu Rev Biophys. 2018 May 20;47:85-106. doi: 10.1146/annurev-biophys-070317-033037. Epub 2018 Jan 18. PMID: 29345990 https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F27313041&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=WltnGFClLoTg0Mwm0wbWCyFQUOyCx27fw0awOdoXSuQ%3D&reserved=0 Translation dynamics of single mRNAs in live cells and neurons. <https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpubmed%2F27313041&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=WltnGFClLoTg0Mwm0wbWCyFQUOyCx27fw0awOdoXSuQ%3D&reserved=0> *Wu B*, Eliscovich C, Yoon YJ,*Singer RH*. Science. 2016 Jun 17;352(6292):1430-5. doi: 10.1126/science.aaf1084. Epub 2016 May 5. PMID: 27313041 Disclosure: Prof. Bin Wu is now my supervisor at JHU (I am writing this on my own time). enjoy, George On 7/19/2018 11:43 AM, Periasamy, Ammasi (ap3t) wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists. > umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7CNi > colai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517d > b44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=EQ1QsQWu > NihXfIP3IoIVS4oVgFb8z9rqmJJiQLEyjCA%3D&reserved=0 > Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=eNVq8k41GUXSOPpTozqymSXH7GaeX6nzd0%2FBMWYtptY%3D&reserved=0 and include the link in your posting. > ***** > > George, > I do not agree with you, it's waste of time to do FRET with fixed cells. As you know there are number of papers published on acceptor photobleaching not only in live cells but also in fixed cells. > For example, RNA-protein interaction there is no way you can work with live cells. > Rehman, S., Gladman, J.T., Periasamy, A., Sun, Y. and Mahadevan, M.S. (2014) Development of an AP-FRET based analysis for characterinzing RNA-protein interactions in myotonic dystrophy (DM1). PLoS ONE 9(4):e95957. > > On FLIM-FRET, we have tested for fixed cells (not published), the lifetime changes only after a week or two of fixation. Yes, FLIM has to be used for live specimens. For some experiment you do not have a choice except to use fixed cells for FRET. > Hope this helps. > Ammasi > > Dr. Ammasi Periasamy > Professor & Center Director, > WM Keck Center for Cellular Imaging, > Department of Biology, Univeristy of Virginia, > 90 Geldard Dr., Charlottesville, VA 22904, USA. > > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.kc > ci.virginia.edu%2Fpeople%2Fprofile%2Fap3t&data=02%7C01%7CNicolai.U > rban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a > 5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=Neg%2Fk4hxDpHfg > ZqjXF%2BHxOK%2Fu1Nbqy5rWnf%2FFqTvM5E%3D&reserved=0 > Phone: (434) 243-7602 or 982-4869 > Fax: (434) 982-5210 > E-mail: [hidden email] > > FRET/FLIM Workshop-March 11-15, 2019: > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.kc > ci.virginia.edu%2Fworkshop&data=02%7C01%7CNicolai.Urban%40MPFI.ORG > %7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0 > %7C0%7C1%7C636676147293414520&sdata=YsGpRRJy3QLTpGltlcgvKla7I1m4T6 > 3sw53MD3kDZ3w%3D&reserved=0 > > > > -----Original Message----- > From: Confocal Microscopy List <[hidden email]> On > Behalf Of Martin Wessendorf > Sent: Thursday, July 19, 2018 10:57 AM > To: [hidden email] > Subject: Re: Weird issue with FRET after acceptor PB > > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists. > umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7CNi > colai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517d > b44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=EQ1QsQWu > NihXfIP3IoIVS4oVgFb8z9rqmJJiQLEyjCA%3D&reserved=0 > Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=eNVq8k41GUXSOPpTozqymSXH7GaeX6nzd0%2FBMWYtptY%3D&reserved=0 and include the link in your posting. > ***** > > Hey, George! > > It'd be helpful if you could offer some references or give your reasoning for thinking that fixed-tissue FRET is a waste of time. > > Thanks! > > Martin Wessendorf > > > > > On 7/19/2018 8:52 AM, George McNamara wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists >> .umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7C >> Nicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b455 >> 17db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293414520&sdata=EQ1Q >> sQWuNihXfIP3IoIVS4oVgFb8z9rqmJJiQLEyjCA%3D&reserved=0 >> Post images on https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293424528&sdata=qmdVJYi0B5JBc1iFbka7UJxbdzn697HSLSCSFJgxl7A%3D&reserved=0 and include the link in your posting. >> ***** >> >> Hi Avi, >> >> FRET of FP pairs after fixation is a complete waste of time (don't >> believe any literature that claims it is useful -- even if authored >> by confocal listserv members). By extension, FRET between fluorescent >> antibodies (Fab's or nanobodies would be smaller, so potentially >> bigger dynamic range) of epitopes detected after fixation may be >> subject to the same issue(s). >> >> The simple test is compare your live vs fixed by acceptor photobleaching. >> >> The best way to do FRET quantitation is FLIM on live cells (and >> again, waste of time to do FLIM-FRET measurements of FP's on fixed cells). >> Traditionally this has been donor FLIM only -- I suggest measuring >> donor and acceptor FLIM simultaneously is better, and several >> companies now offer 30+ million counts per second photon counting >> (see p.s.). >> >> >> George >> >> p.s. disclosure: I may be hosting Leica FALCON (Sept) and ISS >> FastFLIM >> (October?) demonstrations/workshops here at JHU. >> >> I'm also now (as of this email) thinking I should attend the >> "Advanced TCSPC Techniques" workshop in Bethesda Oct 9-10, 2018, that >> Wolfgang Becker recently posted about. >> >> https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww. >> eventbrite.com%2Fe%2F12th-annual-workshop-on-advanced-tcspc-te&da >> ta=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952 >> ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293424528&a >> mp;sdata=ttRqqObOWCjK5ggEHEibEXUK63WXhZS%2B7ojmWqjpmpA%3D&reserve >> d=0 >> chniques-in-the-biomedical-sciences-tickets-46820517428 >> >> >> TCSPC has historically been "painfully slow, small field of view" >> FLIM ( under 1 million photon counts per second ... which would be >> average >> ~1 photon per pixel per second if imaging 1000x1000 pixels, and more >> usefully ~100 counts per pixel per second for 100x100 pixels)). I am >> intrigued by FALCON, FastFLIM, and similar from Becker&Hickl and >> probably others (PicoQuant) because a lot faster (>=30 million counts >> per counting channel - different companies may have different ways to >> up the count rate). >> >> FYI - I am also intrigued by the potential of using "fast counting" >> electronics for simple photon counting at ~30x faster than I can do >> now on our current SP8 HyD's >> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fconfo >> cal.jhu.edu%2Fcurrent-equipment%2Fleica-sp8-confocal-microscop&da >> ta=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952 >> ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293424528&a >> mp;sdata=s6dXLzMAQYm%2FHuYEM2LupsU5g1tREIdsGmQBz%2FONKQw%3D&reser >> ved=0 e (I have not updated FALCON visit dates - yet). That is, not >> -- yet >> -- investing in pulsed lasers and simply 'count faster'. >> >> >> On 7/19/2018 3:33 AM, Avi Jacob wrote: >>> ***** >>> confocal microscopy listserv >>> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flist >>> s.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01% >>> 7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b >>> 45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293424528&sdata= >>> zDhjMDioiI5hlV2A0rqfQ0ni8F63lf0%2F2MMhQjSxD10%3D&reserved=0 >>> Post images on >>> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C1%7C636676147293424528&sdata=qmdVJYi0B5JBc1iFbka7UJxbdzn697HSLSCSFJgxl7A%3D&reserved=0 and include the link in your posting. >>> ***** >>> >>> Hi all, >>> >>> Got a weird issue I cannot explain. >>> >>> We are doing acceptor photobleaching FRET with a CFP-YFP pair, on >>> fixed slides, on Leica widefield with scanner. >>> We are getting great initial bleaching of the YFP acceptor, however, >>> there is an almost immediate recovery of fluorescence in the >>> bleached ROI. >>> This is happening also in the YFP only control. >>> >>> Can this possibly be because of the mounting medium? What else can >>> be a mobile fraction on a fixed slide? >>> The signal is good, has the expected expression pattern. >>> >>> Cheers, >>> Avi >>> >>> -- >>> Avi Jacob, Ph.D. >>> Head of The Kanbar Light Microscopy Unit Bar-Ilan University, >>> Ramat-Gan 5290002, Israel >>> https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Ftiny >>> url.com%2FBIU-Microscopy&data=02%7C01%7CNicolai.Urban%40MPFI.ORG >>> %7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603e >>> d0%7C0%7C0%7C636676147293424528&sdata=EjB1kDzvrUq7thw7T9KGaVnlve >>> o240xM2HUAi73pp48%3D&reserved=0 -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.linkedin.com%2Fin%2Fgeorgemcnamara&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=gM2tnefEcURFBRi6BpgGpafagtDGST4lH3V2JtBoFMw%3D&reserved=0 https://na01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fworks.bepress.com%2Fgmcnamara%2F75&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=AKmokCIl3U4uvU7X4CpaRnvavRmC%2FPzjkU0LxpfuuaI%3D&reserved=0 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fmyncbi%2Fbrowse%2Fcollection%2F44962650&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=AwJ6nZ%2FuFcF23A3sjpEDmCIlIqvKFGj5%2B3tDVcEZuN0%3D&reserved=0 https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fconfocal.jhu.edu&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=rLYezHdyGyDl89i7MU0nZjkskGiWJmvtuRXoa2U72bQ%3D&reserved=0 July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fonlinelibrary.wiley.com%2Fdoi%2F10.1002%2Fcphg.42%2Fabstract&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=oFy8s8J0cOI0%2FeTWMzAgRwas%2FOW4bfqd2sv1Im64mmw%3D&reserved=0 supporting materials direct link is https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fonlinelibrary.wiley.com%2Fdoi%2F10.1002%2Fcphg.42%2Ffull%23hg0404-sec-0023&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293424528&sdata=cbe49t5%2BKGpb5%2B6eQxPLrIOMEELHDi%2BuyqFwipQ9MPg%3D&reserved=0 figures at https://na01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fonlinelibrary.wiley.com%2Fdoi%2F10.1002%2Fcphg.42%2Ffigures&data=02%7C01%7CNicolai.Urban%40MPFI.ORG%7C90c7108baa08405e4cc308d5ed952ab4%7C947b45517db44636a5fd1bdcad603ed0%7C0%7C0%7C636676147293434541&sdata=jVWAMo8EdHHYrPk2it2Y39jq8HYkA1WjOO1%2BZa%2F3gMg%3D&reserved=0 |