Hi all,

Ours ceased working because of yet another card breakdown and we do not
feel it's worth putting another 5k EUR into it. Again. So we decided it's
time.

Did anyone try to convert it to a wide-field fl?

Thanks,
Avi
--
Avi Jacob, Ph.D.
Head of Light Microscopy
The Mina & Everard Goodman Faculty of Life Sciences
Bar-Ilan University, Ramat-Gan 5290002, Israel
Cell: 052-5802544 (call here first), Desk: 972-3-5317647
http://tinyurl.com/BIU-Microscopy
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

Yes, we converted the Zeiss LSM510 Axiovert 200M inverted microscope I
managed at UMiami to widefield (internal motorized Z, manual stage).
Went fine. This was several years ago, so a CCD camera was used.

George


On 7/6/2016 7:31 AM, Avi Jacob wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> Ours ceased working because of yet another card breakdown and we do not
> feel it's worth putting another 5k EUR into it. Again. So we decided it's
> time.
>
> Did anyone try to convert it to a wide-field fl?
>
> Thanks,
> Avi
> --
> Avi Jacob, Ph.D.
> Head of Light Microscopy
> The Mina & Everard Goodman Faculty of Life Sciences
> Bar-Ilan University, Ramat-Gan 5290002, Israel
> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
> http://tinyurl.com/BIU-Microscopy

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We are in the process of converting our 510 to a WF scope with motorized z and motorized stage using Metamorph to drive it. We previously did this with our 5 Live scope (We now have two SP8's - one with MP, FLIM/FRET and 3 color STED, the other with digital light sheet plus a separate GSD system. Life is good!).

Anyone interested in the lasers/laser control box scan head for parts can contact me.

Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George McNamara
Sent: Wednesday, July 06, 2016 8:43 AM
To: [hidden email]
Subject: Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

Yes, we converted the Zeiss LSM510 Axiovert 200M inverted microscope I managed at UMiami to widefield (internal motorized Z, manual stage).
Went fine. This was several years ago, so a CCD camera was used.

George


On 7/6/2016 7:31 AM, Avi Jacob wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> Ours ceased working because of yet another card breakdown and we do
> not feel it's worth putting another 5k EUR into it. Again. So we
> decided it's time.
>
> Did anyone try to convert it to a wide-field fl?
>
> Thanks,
> Avi
> --
> Avi Jacob, Ph.D.
> Head of Light Microscopy
> The Mina & Everard Goodman Faculty of Life Sciences Bar-Ilan
> University, Ramat-Gan 5290002, Israel
> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
> http://tinyurl.com/BIU-Microscopy

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
DamirSudar DamirSudar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

In reply to this post by Avi Jacob
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

Indeed, we also converted one very old 510 into a pretty nice wide-field
scope. It is one based on the Axiovert 100M and I found that
Micro-Manager (https://micro-manager.org/) drives all the components on
that model scope, including the dichroic block mover. Add a decent
camera and good to go. (btw. Axiovision also drives all components on
our scope but M-M gives more flexibility in camera choice).

Best of luck,
- Damir

On 7/6/2016 5:31 AM, Avi Jacob wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all,
>
> Ours ceased working because of yet another card breakdown and we do not
> feel it's worth putting another 5k EUR into it. Again. So we decided it's
> time.
>
> Did anyone try to convert it to a wide-field fl?
>
> Thanks,
> Avi
> --
> Avi Jacob, Ph.D.
> Head of Light Microscopy
> The Mina & Everard Goodman Faculty of Life Sciences
> Bar-Ilan University, Ramat-Gan 5290002, Israel
> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
> http://tinyurl.com/BIU-Microscopy

--
Damir Sudar - Affiliate Scientist
Lawrence Berkeley Natl Laboratory / MBIB
One Cyclotron Road, MS 977, Berkeley, CA 94720, USA
T: 510/486-5346 - F: 510/486-5586 - E: [hidden email]
http://biosciences.lbl.gov/profiles/damir-sudar-2/

Visiting Scientist, Oregon Health & Science University
Avi Jacob Avi Jacob
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Damir,
Did you get  help from reps, or did you manage to do this on your own. What
about connection to the MM computer?
 I have not opned it up yet, but everything seems go through the giant
laser box thing. Is it a matter of just connecting cables to a computer?
Thanks,
Avi

--





On Wed, Jul 6, 2016 at 6:31 PM, Damir Sudar <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Avi,
>
> Indeed, we also converted one very old 510 into a pretty nice wide-field
> scope. It is one based on the Axiovert 100M and I found that Micro-Manager (
> https://micro-manager.org/) drives all the components on that model
> scope, including the dichroic block mover. Add a decent camera and good to
> go. (btw. Axiovision also drives all components on our scope but M-M gives
> more flexibility in camera choice).
>
> Best of luck,
> - Damir
>
> On 7/6/2016 5:31 AM, Avi Jacob wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hi all,
>>
>> Ours ceased working because of yet another card breakdown and we do not
>> feel it's worth putting another 5k EUR into it. Again. So we decided it's
>> time.
>>
>> Did anyone try to convert it to a wide-field fl?
>>
>> Thanks,
>> Avi
>> --
>> Avi Jacob, Ph.D.
>> Head of Light Microscopy
>> The Mina & Everard Goodman Faculty of Life Sciences
>> Bar-Ilan University, Ramat-Gan 5290002, Israel
>> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
>> http://tinyurl.com/BIU-Microscopy
>>
>
> --
> Damir Sudar - Affiliate Scientist
> Lawrence Berkeley Natl Laboratory / MBIB
> One Cyclotron Road, MS 977, Berkeley, CA 94720, USA
> T: 510/486-5346 - F: 510/486-5586 - E: [hidden email]
> http://biosciences.lbl.gov/profiles/damir-sudar-2/
>
> Visiting Scientist, Oregon Health & Science University
>
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

The real time computer (has a bunch of aluminum slats on the front) is
for the confocal components only. Just disconnect from the Axiovert the
scan head and any cables that go to the scanhead and real time computer.

If there is another Aciovert 100 on campus you might compare cabling
with that, and maybe even try Micro-Manager on that first.

George


On 7/10/2016 4:30 AM, Avi Jacob wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Damir,
> Did you get  help from reps, or did you manage to do this on your own. What
> about connection to the MM computer?
>   I have not opned it up yet, but everything seems go through the giant
> laser box thing. Is it a matter of just connecting cables to a computer?
> Thanks,
> Avi
>
> --
>
>
>
>
>
> On Wed, Jul 6, 2016 at 6:31 PM, Damir Sudar <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hi Avi,
>>
>> Indeed, we also converted one very old 510 into a pretty nice wide-field
>> scope. It is one based on the Axiovert 100M and I found that Micro-Manager (
>> https://micro-manager.org/) drives all the components on that model
>> scope, including the dichroic block mover. Add a decent camera and good to
>> go. (btw. Axiovision also drives all components on our scope but M-M gives
>> more flexibility in camera choice).
>>
>> Best of luck,
>> - Damir
>>
>> On 7/6/2016 5:31 AM, Avi Jacob wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your posting.
>>> *****
>>>
>>> Hi all,
>>>
>>> Ours ceased working because of yet another card breakdown and we do not
>>> feel it's worth putting another 5k EUR into it. Again. So we decided it's
>>> time.
>>>
>>> Did anyone try to convert it to a wide-field fl?
>>>
>>> Thanks,
>>> Avi
>>> --
>>> Avi Jacob, Ph.D.
>>> Head of Light Microscopy
>>> The Mina & Everard Goodman Faculty of Life Sciences
>>> Bar-Ilan University, Ramat-Gan 5290002, Israel
>>> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
>>> http://tinyurl.com/BIU-Microscopy
>>>
>> --
>> Damir Sudar - Affiliate Scientist
>> Lawrence Berkeley Natl Laboratory / MBIB
>> One Cyclotron Road, MS 977, Berkeley, CA 94720, USA
>> T: 510/486-5346 - F: 510/486-5586 - E: [hidden email]
>> http://biosciences.lbl.gov/profiles/damir-sudar-2/
>>
>> Visiting Scientist, Oregon Health & Science University
>>

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
DamirSudar DamirSudar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

As George says, the huge box with lasers, power supplies, RTC, etc. all
become surplus and any cables between the Axiovert and that box can be
disconnected. I look briefly on your facility's website and noticed that
your LSM 510 is built on a Axiovert 200M. That's a great scope and very
well supported by MM through the ZeissCAN adapter (not the ZeissCAN29
one!!). All your motorized components will be controlled by MM. A few
things I note from looking at the picture on:
http://lifefaculty.biu.ac.il/FacultyInstruments/index.php?option=com_content&view=article&id=2&Itemid=71
- I cannot see where the META scanhead is connected. Is the right port?
I don't know how to properly defeat the laser interlock on a 200M so you
may need to ask the Zeiss technician for help on that.
- It appears that you don't have a scanning stage but if you do, its
controller (MCU27 or 28) will need to be plugged through a CAN bus cable
into the back of the Axiovert 200M. So, a single RS232 serial cable is
all you'll need between the Axiovert and the MM computer.
- It looks like you have a transmitted PMT. You'll need to take that and
the switcher off of the transmitted light pillar and mount the
transmitted light lamp directly onto the back of the pillar.

Cheers,
- Damir

On 7/12/2016 5:48 AM, George McNamara wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Avi,
>
> The real time computer (has a bunch of aluminum slats on the front) is
> for the confocal components only. Just disconnect from the Axiovert
> the scan head and any cables that go to the scanhead and real time
> computer.
>
> If there is another Aciovert 100 on campus you might compare cabling
> with that, and maybe even try Micro-Manager on that first.
>
> George
>
>
> On 7/10/2016 4:30 AM, Avi Jacob wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your
>> posting.
>> *****
>>
>> Hi Damir,
>> Did you get  help from reps, or did you manage to do this on your
>> own. What
>> about connection to the MM computer?
>>   I have not opned it up yet, but everything seems go through the giant
>> laser box thing. Is it a matter of just connecting cables to a computer?
>> Thanks,
>> Avi
>>
>> --
>>
>>
>>
>>
>>
>> On Wed, Jul 6, 2016 at 6:31 PM, Damir Sudar <[hidden email]> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> Post images on http://www.imgur.com and include the link in your
>>> posting.
>>> *****
>>>
>>> Hi Avi,
>>>
>>> Indeed, we also converted one very old 510 into a pretty nice
>>> wide-field
>>> scope. It is one based on the Axiovert 100M and I found that
>>> Micro-Manager (
>>> https://micro-manager.org/) drives all the components on that model
>>> scope, including the dichroic block mover. Add a decent camera and
>>> good to
>>> go. (btw. Axiovision also drives all components on our scope but M-M
>>> gives
>>> more flexibility in camera choice).
>>>
>>> Best of luck,
>>> - Damir
>>>
>>> On 7/6/2016 5:31 AM, Avi Jacob wrote:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> Post images on http://www.imgur.com and include the link in your
>>>> posting.
>>>> *****
>>>>
>>>> Hi all,
>>>>
>>>> Ours ceased working because of yet another card breakdown and we do
>>>> not
>>>> feel it's worth putting another 5k EUR into it. Again. So we
>>>> decided it's
>>>> time.
>>>>
>>>> Did anyone try to convert it to a wide-field fl?
>>>>
>>>> Thanks,
>>>> Avi
>>>> --
>>>> Avi Jacob, Ph.D.
>>>> Head of Light Microscopy
>>>> The Mina & Everard Goodman Faculty of Life Sciences
>>>> Bar-Ilan University, Ramat-Gan 5290002, Israel
>>>> Cell: 052-5802544 (call here first), Desk: 972-3-5317647
>>>> http://tinyurl.com/BIU-Microscopy
>>>>
>>> --
>>> Damir Sudar - Affiliate Scientist
>>> Lawrence Berkeley Natl Laboratory / MBIB
>>> One Cyclotron Road, MS 977, Berkeley, CA 94720, USA
>>> T: 510/486-5346 - F: 510/486-5586 - E: [hidden email]
>>> http://biosciences.lbl.gov/profiles/damir-sudar-2/
>>>
>>> Visiting Scientist, Oregon Health & Science University
>>>
>

--
Damir Sudar - Affiliate Scientist
Lawrence Berkeley Natl Laboratory / MBIB
One Cyclotron Road, MS 977, Berkeley, CA 94720, USA
T: 510/486-5346 - F: 510/486-5586 - E: [hidden email]
http://biosciences.lbl.gov/profiles/damir-sudar-2/

Visiting Scientist, Oregon Health & Science University
Jan-Peter Siebrasse Jan-Peter Siebrasse
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: What did you do with your old LSM510?

In reply to this post by Avi Jacob
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Avi,

we just did this and as the other replies say, its is easily possible to use the µManager or AxioVision software to control the microscope via the serial port.
HOWEVER, in our case (LSM510 META with Axiovert200M  from 2006) we had to remove an additional safety shutter behind the eyepieces. This shutter could NOT be controlled by any other software but can be easily blocked in open position or completely removed. I did it myself, no need to call the service. If you need more infos/instruction how do do it, you can contact me off list.

Regards, JP

--
**************************************************
Dr. Jan-Peter Siebrasse
[hidden email]

Institut für Physikalische und Theoretische Chemie
AG Biophysikalische Chemie    Universität Bonn
Wegeler Str. 12            53115 Bonn

http://www.thch.uni-bonn.de/pc/kubitscheck/