Why QDs have broad excitation spectrum

classic Classic list List threaded Threaded
♦
♦
4 messages Options
Shalin Mehta Shalin Mehta
Reply | Threaded
Open this post in threaded view
♦
♦
|

Why QDs have broad excitation spectrum

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi All,

 Just curious why QDs don't have narrow absorption spectra like their emission spectra.

For QDs, the energy levels in valence and conduction band of semiconductor atoms are discrete and tunable with size(which lead to emission of photons with certain energy leading to narrow emission spectrum).


During excitation process why photons with any energy (> energy of photons emitted ) can move electrons from valence to conduction band? Doesn't quantization of energy levels play role during excitation?

Cheers
Shalin
--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: electricsbm.blogspot.com
Mobile: +65 90694182
Shalin Mehta Shalin Mehta
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Why QDs have broad excitation spectrum

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I should mention that I think energy levels within valence and conduction bands are also discrete. So as in case of fluorescent dyes, I would expect that photons with only certain energy range can excite electrons in QDs. Instead we see that any sufficiently high energy photon can move electron in QD from valence band to conduction band.


On 9/26/07, Shalin Mehta <[hidden email]> wrote:
Hi All,

 Just curious why QDs don't have narrow absorption spectra like their emission spectra.

For QDs, the energy levels in valence and conduction band of semiconductor atoms are discrete and tunable with size(which lead to emission of photons with certain energy leading to narrow emission spectrum).


During excitation process why photons with any energy (> energy of photons emitted ) can move electrons from valence to conduction band? Doesn't quantization of energy levels play role during excitation?

Cheers
Shalin
--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: <a href="http://electricsbm.blogspot.com" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)"> electricsbm.blogspot.com
Mobile: <span onmouseup="SetCallButtonPressed(this, 0,1)" onmousedown="SetCallButtonPressed(this, 1,1)" onmouseover="SetCallButton(this, 1,1);skype_active=CheckCallButton(this);" onmouseout="SetCallButton(this, 0,1);HideSkypeMenu();" class="skype_tb_injection" id="__skype_highlight_id"> <span style="margin-right: 0px;" title="This is a Singapore phone number. The country code cannot be changed." onclick="javascript:if(0){doRunCMD(event, &#39;chdial&#39;,&#39;0&#39;);}else{doRunCMD(event, &#39;call&#39;,&#39;+6590694182&#39;);}event.preventBubble();return false;" onmouseout="SetCallButtonPart(this, 0);" onmouseover="SetCallButtonPart(this, 1);" class="skype_tb_injection_left" id="__skype_highlight_id_left"> <img src="chrome://skype_ff_toolbar_win/content/cb_transparent_l.gif" style="height: 11px; width: 7px;" class="skype_tb_img_adge" height="11"> <img style="margin: 0px 0px 2px; padding: 0px 1px 1px 0px; width: 16px; top: 0px; left: 0px;" src="chrome://skype_ff_toolbar_win/content/famfamfam/sg.gif" title="" class="skype_tb_img_flag" name="skype_tb_img_f0">
<span title="Call this phone number in Singapore with Skype: +6590694182" onclick="javascript:doRunCMD(event, &#39;call&#39;,&#39;+6590694182&#39;);event.preventBubble();return false;" onmouseout="SetCallButtonPart(this, 0)" onmouseover="SetCallButtonPart(this, 1)" class="skype_tb_injection_right" id="__skype_highlight_id_right"> <img src="chrome://skype_ff_toolbar_win/content/space.gif" style="margin: 0px; padding: 0px; height: 1px; width: 1px;" class="skype_tb_img_space" height="1" width="1"> <img src="chrome://skype_ff_toolbar_win/content/space.gif" style="margin: 0px; padding: 0px; height: 1px; width: 1px;" class="skype_tb_img_space" height="1" width="1"><img src="chrome://skype_ff_toolbar_win/content/space.gif" style="margin: 0px; padding: 0px; height: 1px; width: 1px;" class="skype_tb_img_space" height="1" width="1"> <img src="chrome://skype_ff_toolbar_win/content/space.gif" style="margin: 0px; padding: 0px; height: 1px; width: 1px;" class="skype_tb_img_space" height="1" width="1">+65 90694182 <img src="chrome://skype_ff_toolbar_win/content/cb_transparent_r.gif" style="height: 11px; width: 19px;" class="skype_tb_img_adge" height="11">



--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: electricsbm.blogspot.com
Mobile: +65 90694182
Matt Bootman Matt Bootman
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Why QDs have broad excitation spectrum

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Shalin,

 

When QDs absorb a photon, an electron transfers from the valence band to the conduction band in the semiconductor alloy of the QD core. This creates an exciton (electron-hole pair) that recombines to emit a photon. This happens whenever the energy of the absorbed photon is above the bandgap energy between the two elements (usually Cd and Se) in the core. Excitons can be created over a wide range of energies within the core that correspond to the energy of the absorbed photon. However, the higher-energy excitons always “relax” to the lowest bandgap energy before they recombine and emit a photon. That is why absorbance is broad but emission is narrow.

 

Regards,

 

Matt

Matt Bootman
CEO
Crystalplex Corporation
890 William Pitt Way
Pittsburgh, PA 15238
http://www.crystalplex.com
412.826.3081
412.826.3083 fx
724.344.7128 cell

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Shalin Mehta
Sent: Tuesday, September 25, 2007 8:59 PM
To: [hidden email]
Subject: Re: Why QDs have broad excitation spectrum

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I should mention that I think energy levels within valence and conduction bands are also discrete. So as in case of fluorescent dyes, I would expect that photons with only certain energy range can excite electrons in QDs. Instead we see that any sufficiently high energy photon can move electron in QD from valence band to conduction band.

On 9/26/07, Shalin Mehta <[hidden email]> wrote:

Hi All,

 Just curious why QDs don't have narrow absorption spectra like their emission spectra.

For QDs, the energy levels in valence and conduction band of semiconductor atoms are discrete and tunable with size(which lead to emission of photons with certain energy leading to narrow emission spectrum).


During excitation process why photons with any energy (> energy of photons emitted ) can move electrons from valence to conduction band? Doesn't quantization of energy levels play role during excitation?

Cheers
Shalin
--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: <a href="http://electricsbm.blogspot.com" target="_blank" onclick="return top.js.OpenExtLink(window,event,this)">electricsbm.blogspot.com
Mobile: <span onmouseup="SetCallButtonPressed(this, 0,1)" onmousedown="SetCallButtonPressed(this, 1,1)" onmouseover="SetCallButton(this, 1,1);skype_active=CheckCallButton(this);" onmouseout="SetCallButton(this, 0,1);HideSkypeMenu();" id="__skype_highlight_id"><span title="This is a Singapore phone number. The country code cannot be changed." onclick="javascript:if(0){doRunCMD(event, 'chdial','0');}else{doRunCMD(event, 'call','+6590694182');}event.preventBubble();return false;" onmouseout="SetCallButtonPart(this, 0);" onmouseover="SetCallButtonPart(this, 1);" id="__skype_highlight_id_left"><img border=0 width=11 height=11 src="chrome://skype_ff_toolbar_win/content/cb_transparent_l.gif" style='height:11px;width:7px' class="skype_tb_img_adge"><img border=0 width=32 height=32 src="chrome://skype_ff_toolbar_win/content/famfamfam/sg.gif" style='left:0px; margin-bottom:2px;margin-left:0px;margin-right:0px;margin-top:0px;padding-bottom: 1px;padding-left:0px;padding-right:1px;padding-top:0px;top:0px;width:16px' title="" class="skype_tb_img_flag" name="skype_tb_img_f0">
<span title="Call this phone number in Singapore with Skype: +6590694182" onclick="javascript:doRunCMD(event, 'call','+6590694182');event.preventBubble();return false;" onmouseout="SetCallButtonPart(this, 0)" onmouseover="SetCallButtonPart(this, 1)" id="__skype_highlight_id_right"><img border=0 width=1 height=1 src="chrome://skype_ff_toolbar_win/content/space.gif" style='height:1px;margin-bottom:0px;margin-left:0px;margin-right:0px; margin-top:0px;padding-bottom:0px;padding-left:0px;padding-right:0px; padding-top:0px;width:1px' class="skype_tb_img_space"><img border=0 width=1 height=1 src="chrome://skype_ff_toolbar_win/content/space.gif" style='height:1px;margin-bottom:0px;margin-left:0px;margin-right:0px; margin-top:0px;padding-bottom:0px;padding-left:0px;padding-right:0px; padding-top:0px;width:1px' class="skype_tb_img_space"><img border=0 width=1 height=1 src="chrome://skype_ff_toolbar_win/content/space.gif" style='height:1px;margin-bottom:0px;margin-left:0px;margin-right:0px; margin-top:0px;padding-bottom:0px;padding-left:0px;padding-right:0px; padding-top:0px;width:1px' class="skype_tb_img_space"><img border=0 width=1 height=1 src="chrome://skype_ff_toolbar_win/content/space.gif" style='height:1px;margin-bottom:0px;margin-left:0px;margin-right:0px; margin-top:0px;padding-bottom:0px;padding-left:0px;padding-right:0px; padding-top:0px;width:1px' class="skype_tb_img_space">+65 90694182 <img border=0 width=11 height=11 src="chrome://skype_ff_toolbar_win/content/cb_transparent_r.gif" style='height:11px;width:19px' class="skype_tb_img_adge">




--
My co-ordinates:
Shalin Mehta, Graduate student
Graduate Programme in Bioengineering, NUS, Singapore
Email: shalin {dot} mehta {at} gmail {dot} com
Blog: electricsbm.blogspot.com
Mobile: +65 90694182

John Oreopoulos John Oreopoulos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Labview virtual instrument for Roper cascade 512B EMCCD camera?

In reply to this post by Shalin Mehta
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello listserver,

Does anyone by chance have a Labview VI program that can run a Roper Cascade 512B EMCCD camera? For example, I'm looking for a Labview Vi that would allow me to setup experimental conditions for the camera (ROIs, exposure, temperature, etc.), collect data from the camera, perform online math and display and store the data, etc.

Thank you.


John Oreopoulos, BSc,

PhD Candidate

University of Toronto

Institute For Biomaterials and Biomedical Engineering

Centre For Studies in Molecular Imaging


Tel: W:416-946-5022