Hello All,

I am seeing this problem recently on our LSM510. During a timelapse, the
specimen is going out of focus (drifting down). Initially I thought it maybe
because of the 40X water lens but its happening with the 40X oil and 20x
lens too (so I assume it maybe true with 10X also-not tried that yet).

The scope is equipped with a XL-3 incubator and temp is constant at 37deg. I
have let the Argon laser warm up for 45mins to 1hr before the timelapse and
it still happens. I have tried imaging with 2P chameleon and it still happens.

Anyone experienced this problem before? Is this temperature related or the
focus board on the scope is about to die? Any suggestions appreciated.

Thank you very much.

Abhi



Abhijit Bugde, M.S
Research Associate, Cell Biology
Live Cell Imaging Core Facility
UT Southwestern Medical Center
5323 Harry Hines Blvd.
Dallas, TX 75390-9039

T: 214-648-9935
Michael Schell Michael Schell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Abhijit,

Almost always, drift is caused by a temperature dis-equilibrium.  Can you warm up your XL-3 chamber overnight before imaging?  That usually fixes the drift problem on our XL-3, which does not have a focus board.  Also, we never put the heating unit knob past the middle position.  Takes longer to warm up, but does so more evenly.   Of course, an overnight warmup may not be compatible with a with multi-user core system.  

MIke

On May 16, 2008, at 5:38 PM, Abhijit Bugde wrote:

Search the CONFOCAL archive at

Hello All,

I am seeing this problem recently on our LSM510. During a timelapse, the
specimen is going out of focus (drifting down). Initially I thought it maybe
because of the 40X water lens but its happening with the 40X oil and 20x
lens too (so I assume it maybe true with 10X also-not tried that yet). 

The scope is equipped with a XL-3 incubator and temp is constant at 37deg. I
have let the Argon laser warm up for 45mins to 1hr before the timelapse and
it still happens. I have tried imaging with 2P chameleon and it still happens. 

Anyone experienced this problem before? Is this temperature related or the
focus board on the scope is about to die? Any suggestions appreciated. 

Thank you very much.

Abhi



Abhijit Bugde, M.S
Research Associate, Cell Biology
Live Cell Imaging Core Facility
UT Southwestern Medical Center
5323 Harry Hines Blvd.
Dallas, TX 75390-9039

T: 214-648-9935

Michael J. Schell, Ph.D., CIV, USUHS
Assist. Professor
Dept. of Pharmacology
Uniformed Services University
4301 Jones Bridge Rd.
Bethesda, MD  20814-3220
tel:  (301) 295-3249

Nicolas Tran-Khanh Nicolas Tran-Khanh
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

In reply to this post by Abhijit Bugde
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Abhi,

I had the same problem before. I realized that the wire connected to the stage was too tight in the hole of the incubator, causing strain on the stage and z-drift. Thus, just before starting the timelapse, I make sure that the wire do not produce any pressure on the stage, allowing 24-hours timelapses without any z-drift.

Good luck,

Nicolas

-----Original Message-----
From: Abhijit Bugde [mailto:[hidden email]]
Sent: May 16, 2008 5:38 PM
To: [hidden email]
Subject: Z-Drift-LSM510

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello All,

I am seeing this problem recently on our LSM510. During a timelapse, the
specimen is going out of focus (drifting down). Initially I thought it maybe
because of the 40X water lens but its happening with the 40X oil and 20x
lens too (so I assume it maybe true with 10X also-not tried that yet).

The scope is equipped with a XL-3 incubator and temp is constant at 37deg. I
have let the Argon laser warm up for 45mins to 1hr before the timelapse and
it still happens. I have tried imaging with 2P chameleon and it still happens.

Anyone experienced this problem before? Is this temperature related or the
focus board on the scope is about to die? Any suggestions appreciated.

Thank you very much.

Abhi



Abhijit Bugde, M.S
Research Associate, Cell Biology
Live Cell Imaging Core Facility
UT Southwestern Medical Center
5323 Harry Hines Blvd.
Dallas, TX 75390-9039

T: 214-648-9935
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

In reply to this post by Michael Schell
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal The objective itself is also a tremendous heat sink, although if you haven't seen this drifting problem in the past that may not be your problem.

Craig


On Fri, May 16, 2008 at 3:58 PM, Michael Schell <[hidden email]> wrote:
Abhijit,

Almost always, drift is caused by a temperature dis-equilibrium.  Can you warm up your XL-3 chamber overnight before imaging?  That usually fixes the drift problem on our XL-3, which does not have a focus board.  Also, we never put the heating unit knob past the middle position.  Takes longer to warm up, but does so more evenly.   Of course, an overnight warmup may not be compatible with a with multi-user core system.  

MIke

On May 16, 2008, at 5:38 PM, Abhijit Bugde wrote:

Search the CONFOCAL archive at

Hello All,

I am seeing this problem recently on our LSM510. During a timelapse, the
specimen is going out of focus (drifting down). Initially I thought it maybe
because of the 40X water lens but its happening with the 40X oil and 20x
lens too (so I assume it maybe true with 10X also-not tried that yet). 

The scope is equipped with a XL-3 incubator and temp is constant at 37deg. I
have let the Argon laser warm up for 45mins to 1hr before the timelapse and
it still happens. I have tried imaging with 2P chameleon and it still happens. 

Anyone experienced this problem before? Is this temperature related or the
focus board on the scope is about to die? Any suggestions appreciated. 

Thank you very much.

Abhi



Abhijit Bugde, M.S
Research Associate, Cell Biology
Live Cell Imaging Core Facility
UT Southwestern Medical Center
5323 Harry Hines Blvd.
Dallas, TX 75390-9039

T: 214-648-9935

Michael J. Schell, Ph.D., CIV, USUHS
Assist. Professor
Dept. of Pharmacology
Uniformed Services University
4301 Jones Bridge Rd.
Bethesda, MD  20814-3220
tel:  (301) 295-3249


Abhijit Bugde Abhijit Bugde
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

In reply to this post by Abhijit Bugde
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello All,

Thank you very much for all the replies. I will try to answer most of the
questions in 1 post.

-I haven’t checked if the wires are causing any strain on the stage. Hope
it’s as simple as that for me too.

-We are using the EXFO lamp so it’s not heating the stand.

-I have kept the incubator at a steady temp overnight but that does not
solve the drift problem.

-Always used the turret loaded with objectives but will try to image with 1
objective to find out if its gravity causing the problem.

Again, thank you for the feedback.

Best regards,
Abhi
Andreas V Andreas V
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been struggling with the same problem for a long time. End up
correcting the shift manually between imaging points, since there was always
a drift in same direction (approximately 2-3µm/h). Not the most elegant way
but it worked out well.

Could not find another solution until now (unfortunately including
suggestions that were given here already).

There might be a software solution for the Zeiss imaging system at
http://www.zeiss.de/lsm-macros (see multiple time series) I have never tried
it though.

Good luck,
Andreas

>----------------------------------------------
> Andreas Vlachos
> Neuroscience Center
> Institute of Clinical Neuroanatomy
> Goethe-University Frankfurt
> Heinrich-Hofmann street  7
> D-60590 Frankfurt am Main, Germany
> Phone: ++49  (69) 6301 83323



 





> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Abhijit Bugde
> Sent: Samstag, 17. Mai 2008 05:27
> To: [hidden email]
> Subject: Re: Z-Drift-LSM510
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hello All,
>
> Thank you very much for all the replies. I will try to answer most of
> the
> questions in 1 post.
>
> -I haven’t checked if the wires are causing any strain on the stage.
> Hope
> it’s as simple as that for me too.
>
> -We are using the EXFO lamp so it’s not heating the stand.
>
> -I have kept the incubator at a steady temp overnight but that does not
> solve the drift problem.
>
> -Always used the turret loaded with objectives but will try to image
> with 1
> objective to find out if its gravity causing the problem.
>
> Again, thank you for the feedback.
>
> Best regards,
> Abhi
Hugo.Ostermann Hugo.Ostermann
Reply | Threaded
Open this post in threaded view
♦
♦
|

AW: Z-Drift-LSM510

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Have you tried to remove all other objectives? We found that cures the
problem on our Applied Precision DeltaVision Systems which are based on the
IX-71 OLYPUS. It also has a tension ring which we set to strong.

 
Good luck

Hugo Ostermann
 
CHROMAPHOR Analysen-Technik GmbH
Mühlenkamp 37
D-59387 Ascheberg - Germany
Tel:   ++49 - 2593-928.600
Fax:  ++49 - 2593-928.601
mail:  [hidden email]
URL: http://www.chromaphor.de
Sitz Ascheberg-Westfalen
Amtsgericht - Registergericht Coesfeld HRB 7703
Geschäftsführer: Hugo Ostermann
 
-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Andreas V
Gesendet: Samstag, 17. Mai 2008 11:53
An: [hidden email]
Betreff: Re: Z-Drift-LSM510

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been struggling with the same problem for a long time. End up
correcting the shift manually between imaging points, since there was always
a drift in same direction (approximately 2-3µm/h). Not the most elegant way
but it worked out well.

Could not find another solution until now (unfortunately including
suggestions that were given here already).

There might be a software solution for the Zeiss imaging system at
http://www.zeiss.de/lsm-macros (see multiple time series) I have never tried
it though.

Good luck,
Andreas

>----------------------------------------------
> Andreas Vlachos
> Neuroscience Center
> Institute of Clinical Neuroanatomy
> Goethe-University Frankfurt
> Heinrich-Hofmann street  7
> D-60590 Frankfurt am Main, Germany
> Phone: ++49  (69) 6301 83323



 





> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Abhijit Bugde
> Sent: Samstag, 17. Mai 2008 05:27
> To: [hidden email]
> Subject: Re: Z-Drift-LSM510
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hello All,
>
> Thank you very much for all the replies. I will try to answer most of
> the
> questions in 1 post.
>
> -I haven’t checked if the wires are causing any strain on the stage.
> Hope
> it’s as simple as that for me too.
>
> -We are using the EXFO lamp so it’s not heating the stand.
>
> -I have kept the incubator at a steady temp overnight but that does not
> solve the drift problem.
>
> -Always used the turret loaded with objectives but will try to image
> with 1
> objective to find out if its gravity causing the problem.
>
> Again, thank you for the feedback.
>
> Best regards,
> Abhi
Beat Ludin Beat Ludin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

In reply to this post by Abhijit Bugde
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Abhi -

The XL3 incubator is only enclosing part of the
microscope and metal has a high heat conductance,
so changes in room temperature may still cause
focus drift. (That's why we design our incubators
to enclose as much of the microscope as possible
=> attention, commercial interest). Monitor the
ambient temperature to see if there is a correlation with focus instability.
That said, I want to add that there are also
other reasons for focus instability:
I presume that you have already checked that the
temperature inside the incubator is really
stable. Fluctuations as small as 0.3°C can cause
clearly noticeable focus variations.
Then keep in mind that not just the microscope
but also the sample needs to be thermally
equilibrated. That can take quite some time, too
(up to several hours, depending on the sample format).
With today's research-grade microscopes,
mechanical slip is much less common than with
their predecessors, but it is still worth
checking for. A good indicator is a monotonous drift following gravity.
With open "elastic" sample carriers (e.g. plastic
dishes), weight reduction due to evaporation of
culture medium can have the opposite effect, i.e.
monotonous drift against gravity. If that is the
case, you should see no drift with fixed embedded samples.
With piezo or galvo Z-drives, electrical drift is
also a candidate. Check if you still see drift if
you de-power or remove the device.
Last but not least, variations in heat
dissipation of electrical devices (e.g. a
motorized stage) can cause local thermal
fluctuations that cannot be compensated by the
incubator. Again, de-power the device in question
to check if it is the culprit.

Best regards,

Beat
Life Imaging Services


At 05:26 17-05-2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Hello All,
>
>Thank you very much for all the replies. I will try to answer most of the
>questions in 1 post.
>
>-I haven’t checked if the wires are causing any strain on the stage. Hope
>it’s as simple as that for me too.
>
>-We are using the EXFO lamp so it’s not heating the stand.
>
>-I have kept the incubator at a steady temp overnight but that does not
>solve the drift problem.
>
>-Always used the turret loaded with objectives but will try to image with 1
>objective to find out if its gravity causing the problem.
>
>Again, thank you for the feedback.
>
>Best regards,
>Abhi
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Z-Drift-LSM510

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Abhi,

The advices given so far assume that the drift is not caused by a real displacement of the Z motor of your scope. However, I recently had the problem of a faulty encoder inside the Z knob of the touchscreen remote control unit of a Zeiss Axio Observer, causing the focus to change in a non-reproducible maner (sometimes as bad as 1 micron/minute). The ironic part is that I spent some time checking temperature, vibration, etc... before realizing that the drift was actually monitored by the Z motor position, indicating a motor problem. So far the Zeiss people have changed the touchscreen unit two times, and the interface card inside the scope, and the problem is attenuated but still present. The solutions are to record the Z position and force to re-actuate it before each images, or just unplug the touchscreen unit before launching a timelapse (altough it may be not so good to "hot" unplug the unit).

So I'd suggest you check first if the drift is recorded by your stage/Z motor position (on the scope or inside the software you use), and if re-actuating the Z-position to a recorded point can compensate for the drift. If not, then the problems are thermal instability or mechanical problems such as vibrations as people pointed out previouly.

Christophe




On Sat, May 17, 2008 at 9:23 PM, Beat Ludin <[hidden email]> wrote:
Abhi -

The XL3 incubator is only enclosing part of the microscope and metal has a high heat conductance, so changes in room temperature may still cause focus drift. (That's why we design our incubators to enclose as much of the microscope as possible => attention, commercial interest). Monitor the ambient temperature to see if there is a correlation with focus instability.
That said, I want to add that there are also other reasons for focus instability:
I presume that you have already checked that the temperature inside the incubator is really stable. Fluctuations as small as 0.3°C can cause clearly noticeable focus variations.
Then keep in mind that not just the microscope but also the sample needs to be thermally equilibrated. That can take quite some time, too (up to several hours, depending on the sample format).
With today's research-grade microscopes, mechanical slip is much less common than with their predecessors, but it is still worth checking for. A good indicator is a monotonous drift following gravity.
With open "elastic" sample carriers (e.g. plastic dishes), weight reduction due to evaporation of culture medium can have the opposite effect, i.e. monotonous drift against gravity. If that is the case, you should see no drift with fixed embedded samples.
With piezo or galvo Z-drives, electrical drift is also a candidate. Check if you still see drift if you de-power or remove the device.
Last but not least, variations in heat dissipation of electrical devices (e.g. a motorized stage) can cause local thermal fluctuations that cannot be compensated by the incubator. Again, de-power the device in question to check if it is the culprit.

Best regards,

Beat
Life Imaging Services



At 05:26 17-05-2008, you wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello All,

Thank you very much for all the replies. I will try to answer most of the
questions in 1 post.

-I haven't checked if the wires are causing any strain on the stage. Hope
it's as simple as that for me too.

-We are using the EXFO lamp so it's not heating the stand.

-I have kept the incubator at a steady temp overnight but that does not
solve the drift problem.

-Always used the turret loaded with objectives but will try to image with 1
objective to find out if its gravity causing the problem.

Again, thank you for the feedback.

Best regards,
Abhi