Zeiss SD

classic Classic list List threaded Threaded
♦
♦
10 messages Options
simon walker (BI) simon walker (BI)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Zeiss SD

I was interested to see that Zeiss have gone into partnership with Yokogawa
to launch their own flavour of the spinning disk confocal:
http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
spinning-disk-cell-observer-sd-2008

Does this represent an acknowledgment that the 5-live doesn't do everything
it says on the tin, or just that it's too expensive?  Are there any happy 5-live
users out there?
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
-Esteban

On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <[hidden email]> wrote:
I was interested to see that Zeiss have gone into partnership with Yokogawa
to launch their own flavour of the spinning disk confocal:
http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
spinning-disk-cell-observer-sd-2008


Does this represent an acknowledgment that the 5-live doesn't do everything
it says on the tin, or just that it's too expensive?  Are there any happy 5-live
users out there?



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax
Fred Mast Fred Mast
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration...
By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


Cheers,

Fred

On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
-Esteban

On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <[hidden email]> wrote:
I was interested to see that Zeiss have gone into partnership with Yokogawa
to launch their own flavour of the spinning disk confocal:
http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
spinning-disk-cell-observer-sd-2008


Does this represent an acknowledgment that the 5-live doesn't do everything
it says on the tin, or just that it's too expensive?  Are there any happy 5-live
users out there?



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax

Fred D. Mast
Department of Cell Biology
Medical Sciences Building Room 5-14
University of Alberta
Edmonton, Alberta, T6G 2H7
Canada

Tel: 1-780-492-7407



Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

the Yoko head can be ordered with a choice of disk to match your  
primary lens.
I was told by Zess the licensing agreement restricts them to selling  
the Yokogawa head coupled to an inverted stand with a stage enclosure  
environmental chamber.


Regards,
Glen
On Feb 10, 2009, at 11:10 AM, Fred Mast wrote:

> Does the CSU-X1 allow you to change the disks? I'm still waiting for  
> a spinning-disk unit that is actually designed for live-cell  
> imaging. The previous Yokogawa units were optimized for 100x 1,4 NA  
> oil objectives if I'm not mistaken. Even with relatively thin  
> samples (cultured cells), you're still introducing spherical  
> aberration...
> By comparison, has anyone else looked at the technology being  
> developed in Raphael Yuste's laboratory? They have combined  
> multiphoton LSM with spatial light modulation technology to  
> effectively "scan" an entire field at once. I'm not an expert in  
> this but in my opinion if you could make it fast enough this seems  
> like a better option to me (they report 60Hz calcium ion imaging).  
> You can check it out here (under the SLM Microscopy section):
>
> http://www.columbia.edu/cu/biology/faculty/yuste/methods.html
>
> Cheers,
>
> Fred
>
> On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:
>
>> I too was intrigued when I saw Zeiss roll out a spinning disk.  
>> We're happy with our 5 LIVE.  Being able to vary the slit aperture  
>> size to match different objectives/samples is important in our  
>> multi-user facility.  It's also nice to capture two fluorescent  
>> channels simultaneously.  I wish the detector on the 5 LIVE were EM-
>> CCD but the regular CCD does the job.  Images sometimes have  
>> vertical line artifacts but those are nicely dealt with by Fourier  
>> filtering or dividing by a blank image.  Must be lower cost (if  
>> that's true) and/or established familiarity of the spinning disk  
>> that makes them easy to sell.
>>
>> -Esteban
>>
>> On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <[hidden email]
>> > wrote:
>> I was interested to see that Zeiss have gone into partnership with  
>> Yokogawa
>> to launch their own flavour of the spinning disk confocal:
>> http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
>> spinning-disk-cell-observer-sd-2008
>>
>> Does this represent an acknowledgment that the 5-live doesn't do  
>> everything
>> it says on the tin, or just that it's too expensive?  Are there any  
>> happy 5-live
>> users out there?
>>
>>
>>
>> --
>> G. Esteban Fernandez, Ph.D.
>> Associate Director
>> Molecular Cytology Core Facility
>> University of Missouri
>> 120 Bond Life Sciences Center
>> Columbia, MO  65211
>>
>> http://www.biotech.missouri.edu/mcc/
>>
>> 573-882-4895
>> 573-884-9395 fax
>
> Fred D. Mast
> Department of Cell Biology
> Medical Sciences Building Room 5-14
> University of Alberta
> Edmonton, Alberta, T6G 2H7
> Canada
>
> Tel: 1-780-492-7407
> [hidden email]
>
>
>
RICHARD BURRY RICHARD BURRY
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

In reply to this post by Fred Mast
Fred
We have a Visitech Infinity3 2D array scanning confocal.  It is based on the same principle as the spinning disk but use an array of pin holes.  This confocal allows seven different sizes of pin holes with diameter from 10 um to 60 um.  Our experience is that is gives great performance with a Hamamatsu EM13.  With the 100x 1.4 N.A. the optimal pin hole is 30 um but at 60 um the optical section thickness is a little more than a micron but the intensity is great for even weak fluorescence. 
Dick

----- Original Message -----
From: Fred Mast <[hidden email]>
Date: Tuesday, February 10, 2009 2:12 pm
Subject: Re: Zeiss SD
To: [hidden email]


> Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration... By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


> http://www.columbia.edu/cu/biology/faculty/yuste/methods.html


> Cheers,


> Fred

> On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

> I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
> -Esteban


> On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <<A href="javascript:main.compose('new','t=simon.walker@bbsrc.ac.uk')">simon.walker@...> wrote:
> I was interested to see that Zeiss have gone into partnership with Yokogawa
> to launch their own flavour of the spinning disk confocal:
> http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
> spinning-disk-cell-observer-sd-2008


> Does this represent an acknowledgment that the 5-live doesn't do everything
> it says on the tin, or just that it's too expensive?  Are there any happy 5-live
> users out there?




> --
> G. Esteban Fernandez, Ph.D.
> Associate Director
> Molecular Cytology Core Facility
> University of Missouri
> 120 Bond Life Sciences Center
> Columbia, MO  65211

> http://www.biotech.missouri.edu/mcc/

> 573-882-4895
> 573-884-9395 fax


> Fred D. Mast
> Department of Cell Biology
> Medical Sciences Building Room 5-14
> University of Alberta
> Edmonton, Alberta, T6G 2H7
> Canada


> Tel: 1-780-492-7407
<A href="javascript:main.compose('new','t=fmast@ualberta.ca')">> fmast@...









> Spam
> Not spam
> Forget previous vote



Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849

Watkins, Simon C Watkins, Simon C
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

I tend to agree with Dick here,  The pinhole array solutions while less "mature" than the spinning disk confocals work very well.  We have had a Yokagawa head for ages, and the limitation to a 1.4NA objective was significant.  About a year ago we bought a Prairie swept field array scanning confocal.  It allows multiple pinholes or slits sizes (7)  depending on your specimen and optics used.  It works very well indeed, and in our hands is more flexible than the Yokogawa head.  Its important to note that the yokogawa head I have is very old and there have been many advances.  However without rapid and simple changing of pinhole size (as is possible with the Prairie or Visitech solutions) your image will be compromised when NA is changed.

S.

 

Simon C. Watkins Ph.D, FRCPath

Professor and Vice Chair, Cell Biology and Physiology

Professor, Immunology

Director, Center for Biologic Imaging

BSTS 225, University of Pittsburgh

3500 Terrace St.

Pittsburgh PA 15261

Tel: 412-352-2277

Fax:412-648-2797

URL: http://www.cbi.pitt.edu

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: Tuesday, February 10, 2009 3:20 PM
To: [hidden email]
Subject: Re: Zeiss SD

 

Fred
We have a Visitech Infinity3 2D array scanning confocal.  It is based on the same principle as the spinning disk but use an array of pin holes.  This confocal allows seven different sizes of pin holes with diameter from 10 um to 60 um.  Our experience is that is gives great performance with a Hamamatsu EM13.  With the 100x 1.4 N.A. the optimal pin hole is 30 um but at 60 um the optical section thickness is a little more than a micron but the intensity is great for even weak fluorescence. 
Dick

----- Original Message -----
From: Fred Mast <[hidden email]>
Date: Tuesday, February 10, 2009 2:12 pm
Subject: Re: Zeiss SD
To: [hidden email]

> Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration... By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


> http://www.columbia.edu/cu/biology/faculty/yuste/methods.html


> Cheers,


> Fred

> On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

> I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
> -Esteban

> On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <<a href="javascript:main.compose('new','t=simon.walker@bbsrc.ac.uk')">simon.walker@...> wrote:

> I was interested to see that Zeiss have gone into partnership with Yokogawa
> to launch their own flavour of the spinning disk confocal:
> http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
> spinning-disk-cell-observer-sd-2008


> Does this represent an acknowledgment that the 5-live doesn't do everything
> it says on the tin, or just that it's too expensive?  Are there any happy 5-live
> users out there?





> --
> G. Esteban Fernandez, Ph.D.
> Associate Director
> Molecular Cytology Core Facility
> University of Missouri
> 120 Bond Life Sciences Center
> Columbia, MO  65211

> http://www.biotech.missouri.edu/mcc/

> 573-882-4895
> 573-884-9395 fax

 

> Fred D. Mast
> Department of Cell Biology
> Medical Sciences Building Room 5-14
> University of Alberta
> Edmonton, Alberta, T6G 2H7
> Canada


> Tel: 1-780-492-7407
<a href="javascript:main.compose('new','t=fmast@ualberta.ca')">> fmast@...








> Spam
> Not spam
> Forget previous vote



Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849

Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

Presumably the problem is that the Yokogawa is designed for x60 NA 1.4 -
it's the magnification as well as the NA that determines the size of the Airy
disk at the pinhole.  So a x20 NA 0.5, which is a common lens, ought to
work pretty well (one third the mag and ~ one third the NA).
 
                                                                                Guy
 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C
Sent: Wednesday, 11 February 2009 7:47 AM
To: [hidden email]
Subject: Re: Zeiss SD

I tend to agree with Dick here,  The pinhole array solutions while less "mature" than the spinning disk confocals work very well.  We have had a Yokagawa head for ages, and the limitation to a 1.4NA objective was significant.  About a year ago we bought a Prairie swept field array scanning confocal.  It allows multiple pinholes or slits sizes (7)  depending on your specimen and optics used.  It works very well indeed, and in our hands is more flexible than the Yokogawa head.  Its important to note that the yokogawa head I have is very old and there have been many advances.  However without rapid and simple changing of pinhole size (as is possible with the Prairie or Visitech solutions) your image will be compromised when NA is changed.

S.

 

Simon C. Watkins Ph.D, FRCPath

Professor and Vice Chair, Cell Biology and Physiology

Professor, Immunology

Director, Center for Biologic Imaging

BSTS 225, University of Pittsburgh

3500 Terrace St.

Pittsburgh PA 15261

Tel: 412-352-2277

Fax:412-648-2797

URL: http://www.cbi.pitt.edu

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: Tuesday, February 10, 2009 3:20 PM
To: [hidden email]
Subject: Re: Zeiss SD

 

Fred
We have a Visitech Infinity3 2D array scanning confocal.  It is based on the same principle as the spinning disk but use an array of pin holes.  This confocal allows seven different sizes of pin holes with diameter from 10 um to 60 um.  Our experience is that is gives great performance with a Hamamatsu EM13.  With the 100x 1.4 N.A. the optimal pin hole is 30 um but at 60 um the optical section thickness is a little more than a micron but the intensity is great for even weak fluorescence. 
Dick

----- Original Message -----
From: Fred Mast <[hidden email]>
Date: Tuesday, February 10, 2009 2:12 pm
Subject: Re: Zeiss SD
To: [hidden email]

> Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration... By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


> http://www.columbia.edu/cu/biology/faculty/yuste/methods.html


> Cheers,


> Fred

> On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

> I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
> -Esteban

> On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <<A href="javascript:main.compose('new','t=simon.walker@bbsrc.ac.uk')">simon.walker@...> wrote:

> I was interested to see that Zeiss have gone into partnership with Yokogawa
> to launch their own flavour of the spinning disk confocal:
> http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
> spinning-disk-cell-observer-sd-2008


> Does this represent an acknowledgment that the 5-live doesn't do everything
> it says on the tin, or just that it's too expensive?  Are there any happy 5-live
> users out there?





> --
> G. Esteban Fernandez, Ph.D.
> Associate Director
> Molecular Cytology Core Facility
> University of Missouri
> 120 Bond Life Sciences Center
> Columbia, MO  65211

> http://www.biotech.missouri.edu/mcc/

> 573-882-4895
> 573-884-9395 fax

 

> Fred D. Mast
> Department of Cell Biology
> Medical Sciences Building Room 5-14
> University of Alberta
> Edmonton, Alberta, T6G 2H7
> Canada


> Tel: 1-780-492-7407
<A href="javascript:main.compose('new','t=fmast@ualberta.ca')">> fmast@...








> Spam
> Not spam
> Forget previous vote



Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849


No virus found in this incoming message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM


No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM

RICHARD BURRY RICHARD BURRY
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

Guy
The key is the thickness of the optical section.  While a 100x 1.4 N.A. objective with a 30 um pin hole has an optical section thickness of 0.7 um, a 20x 0.45 N.A. objective has an optical section thickness of 6.4 um.  So you loose a lot with a 20x not only in optical section thickness but also in through put (because of the lower N.A).
Dick
 
Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849

----- Original Message -----
From: Guy Cox <[hidden email]>
Date: Wednesday, February 11, 2009 5:44 am
Subject: Re: Zeiss SD
To: [hidden email]

> Presumably the problem is that the Yokogawa is designed for x60 NA 1.4 -
> it's the magnification as well as the NA that determines the size of the Airy
> disk at the pinhole.  So a x20 NA 0.5, which is a common lens, ought to
> work pretty well (one third the mag and ~ one third the NA).
 
>                                                                                 Guy
 

> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net

 


> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C
> Sent: Wednesday, 11 February 2009 7:47 AM
> To: [hidden email]
> Subject: Re: Zeiss SD



> I tend to agree with Dick here,  The pinhole array solutions while less "mature" than the spinning disk confocals work very well.  We have had a Yokagawa head for ages, and the limitation to a 1.4NA objective was significant.  About a year ago we bought a Prairie swept field array scanning confocal.  It allows multiple pinholes or slits sizes (7)  depending on your specimen and optics used.  It works very well indeed, and in our hands is more flexible than the Yokogawa head.  Its important to note that the yokogawa head I have is very old and there have been many advances.  However without rapid and simple changing of pinhole size (as is possible with the Prairie or Visitech solutions) your image will be compromised when NA is changed.

> S.

 

> Simon C. Watkins Ph.D, FRCPath

> Professor and Vice Chair, Cell Biology and Physiology

> Professor, Immunology

> Director, Center for Biologic Imaging

> BSTS 225, University of Pittsburgh

> 3500 Terrace St.

> Pittsburgh PA 15261

> Tel: 412-352-2277

> Fax:412-648-2797

> URL: http://www.cbi.pitt.edu

 

> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
> Sent: Tuesday, February 10, 2009 3:20 PM
> To: [hidden email]
> Subject: Re: Zeiss SD


 

> Fred
> We have a Visitech Infinity3 2D array scanning confocal.  It is based on the same principle as the spinning disk but use an array of pin holes.  This confocal allows seven different sizes of pin holes with diameter from 10 um to 60 um.  Our experience is that is gives great performance with a Hamamatsu EM13.  With the 100x 1.4 N.A. the optimal pin hole is 30 um but at 60 um the optical section thickness is a little more than a micron but the intensity is great for even weak fluorescence. 
> Dick

> ----- Original Message -----
> From: Fred Mast <[hidden email]>
> Date: Tuesday, February 10, 2009 2:12 pm
> Subject: Re: Zeiss SD
> To: [hidden email]

> > Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration... By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


> > http://www.columbia.edu/cu/biology/faculty/yuste/methods.html


> > Cheers,


> > Fred

> > On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

> I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
> > -Esteban

> > On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <simon.walker@...> wrote:

> > I was interested to see that Zeiss have gone into partnership with Yokogawa
> > to launch their own flavour of the spinning disk confocal:
> > http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
> > spinning-disk-cell-observer-sd-2008


> > Does this represent an acknowledgment that the 5-live doesn't do everything
> > it says on the tin, or just that it's too expensive?  Are there any happy 5-live
> > users out there?





> > --
> > G. Esteban Fernandez, Ph.D.
> > Associate Director
> > Molecular Cytology Core Facility
> > University of Missouri
> > 120 Bond Life Sciences Center
> > Columbia, MO  65211

> > http://www.biotech.missouri.edu/mcc/

> > 573-882-4895
> > 573-884-9395 fax


 

> > Fred D. Mast
> > Department of Cell Biology
> > Medical Sciences Building Room 5-14
> > University of Alberta
> > Edmonton, Alberta, T6G 2H7
> > Canada


> > Tel: 1-780-492-7407
> > fmast@...








> Richard W. Burry, Ph.D.
> Department of Neuroscience, College of Medicine
> Campus Microscopy and Imaging Facility, Director
> The Ohio State University
> Associate Editor, Journal of Histochemistry and Cytochemistry
> 277 Biomedical Research Tower
> 460 West Twelfth Avenue
> Columbus, Ohio 43210
> Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849



> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM



> Spam
> Not spam
> Forget previous vote


> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM





Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

In reply to this post by simon walker (BI)
There is also technology called Programmable Array Microscopy which was
developed in the lab of Tom Jovin. Cairn Research Limited was beta testing
this about 2 years ago:

http://www.cairnweb.com/newsletter/pam.html

However I could not find any updates on their new web site:

http://www.cairn-research.co.uk/Cairn

Perhaps others on the list know where things are with PAM.

Basically it is a programmable array so  you can open as many local array
elements to essentially vary the size of the pinhole. You can also simulate
any pattern you like so you could make it perform like swept field or like a
spinning disk and so on. One of the points I understood is you can also
collect the out of focus light for use in deconvolution unlike other
systems. I'm not sure how this is done though.

It sounds like there is a lot of promise but as I said I'm not sure where
things are  now. There are also primary research articles out of the Jovin lab.
Sincerely,

Claire
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Zeiss SD

In reply to this post by RICHARD BURRY
Sure, but that applies to any confocal - optical section thickness
goes as the inverse square of the NA.  By and large if you want
optimal optical sectioning you won't use a spinning disk anyway.
My point was that there will almost always be more than one lens
which matches a given pinhole size.  
 
                                                                           Guy 
 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: Thursday, 12 February 2009 12:20 AM
To: [hidden email]
Subject: Re: Zeiss SD

Guy
The key is the thickness of the optical section.  While a 100x 1.4 N.A. objective with a 30 um pin hole has an optical section thickness of 0.7 um, a 20x 0.45 N.A. objective has an optical section thickness of 6.4 um.  So you loose a lot with a 20x not only in optical section thickness but also in through put (because of the lower N.A).
Dick
 
Richard W. Burry, Ph.D.
Department of Neuroscience, College of Medicine
Campus Microscopy and Imaging Facility, Director
The Ohio State University
Associate Editor, Journal of Histochemistry and Cytochemistry
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210
Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849

----- Original Message -----
From: Guy Cox <[hidden email]>
Date: Wednesday, February 11, 2009 5:44 am
Subject: Re: Zeiss SD
To: [hidden email]

> Presumably the problem is that the Yokogawa is designed for x60 NA 1.4 -
> it's the magnification as well as the NA that determines the size of the Airy
> disk at the pinhole.  So a x20 NA 0.5, which is a common lens, ought to
> work pretty well (one third the mag and ~ one third the NA).

>                                                                                 Guy
 

> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net

 


> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C
> Sent: Wednesday, 11 February 2009 7:47 AM
> To: [hidden email]
> Subject: Re: Zeiss SD



> I tend to agree with Dick here,  The pinhole array solutions while less "mature" than the spinning disk confocals work very well.  We have had a Yokagawa head for ages, and the limitation to a 1.4NA objective was significant.  About a year ago we bought a Prairie swept field array scanning confocal.  It allows multiple pinholes or slits sizes (7)  depending on your specimen and optics used.  It works very well indeed, and in our hands is more flexible than the Yokogawa head.  Its important to note that the yokogawa head I have is very old and there have been many advances.  However without rapid and simple changing of pinhole size (as is possible with the Prairie or Visitech solutions) your image will be compromised when NA is changed.

> S.

 

> Simon C. Watkins Ph.D, FRCPath

> Professor and Vice Chair, Cell Biology and Physiology

> Professor, Immunology

> Director, Center for Biologic Imaging

> BSTS 225, University of Pittsburgh

> 3500 Terrace St.

> Pittsburgh PA 15261

> Tel: 412-352-2277

> Fax:412-648-2797

> URL: http://www.cbi.pitt.edu

 

> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
> Sent: Tuesday, February 10, 2009 3:20 PM
> To: [hidden email]
> Subject: Re: Zeiss SD


 

> Fred
> We have a Visitech Infinity3 2D array scanning confocal.  It is based on the same principle as the spinning disk but use an array of pin holes.  This confocal allows seven different sizes of pin holes with diameter from 10 um to 60 um.  Our experience is that is gives great performance with a Hamamatsu EM13.  With the 100x 1.4 N.A. the optimal pin hole is 30 um but at 60 um the optical section thickness is a little more than a micron but the intensity is great for even weak fluorescence. 
> Dick

> ----- Original Message -----
> From: Fred Mast <[hidden email]>
> Date: Tuesday, February 10, 2009 2:12 pm
> Subject: Re: Zeiss SD
> To: [hidden email]

> > Does the CSU-X1 allow you to change the disks? I'm still waiting for a spinning-disk unit that is actually designed for live-cell imaging. The previous Yokogawa units were optimized for 100x 1,4 NA oil objectives if I'm not mistaken. Even with relatively thin samples (cultured cells), you're still introducing spherical aberration... By comparison, has anyone else looked at the technology being developed in Raphael Yuste's laboratory? They have combined multiphoton LSM with spatial light modulation technology to effectively "scan" an entire field at once. I'm not an expert in this but in my opinion if you could make it fast enough this seems like a better option to me (they report 60Hz calcium ion imaging). You can check it out here (under the SLM Microscopy section):


> > http://www.columbia.edu/cu/biology/faculty/yuste/methods.html


> > Cheers,


> > Fred

> > On 10-Feb-09, at 10:55 AM, G. Esteban Fernandez wrote:

> I too was intrigued when I saw Zeiss roll out a spinning disk.  We're happy with our 5 LIVE.  Being able to vary the slit aperture size to match different objectives/samples is important in our multi-user facility.  It's also nice to capture two fluorescent channels simultaneously.  I wish the detector on the 5 LIVE were EM-CCD but the regular CCD does the job.  Images sometimes have vertical line artifacts but those are nicely dealt with by Fourier filtering or dividing by a blank image.  Must be lower cost (if that's true) and/or established familiarity of the spinning disk that makes them easy to sell.
 
> > -Esteban

> > On Tue, Feb 10, 2009 at 10:35 AM, Simon Walker <simon.walker@...> wrote:

> > I was interested to see that Zeiss have gone into partnership with Yokogawa
> > to launch their own flavour of the spinning disk confocal:
> > http://www.microscopy-analysis.com/news/carl-zeiss-and-yokogawa-launch-
> > spinning-disk-cell-observer-sd-2008


> > Does this represent an acknowledgment that the 5-live doesn't do everything
> > it says on the tin, or just that it's too expensive?  Are there any happy 5-live
> > users out there?





> > --
> > G. Esteban Fernandez, Ph.D.
> > Associate Director
> > Molecular Cytology Core Facility
> > University of Missouri
> > 120 Bond Life Sciences Center
> > Columbia, MO  65211

> > http://www.biotech.missouri.edu/mcc/

> > 573-882-4895
> > 573-884-9395 fax


 

> > Fred D. Mast
> > Department of Cell Biology
> > Medical Sciences Building Room 5-14
> > University of Alberta
> > Edmonton, Alberta, T6G 2H7
> > Canada


> > Tel: 1-780-492-7407
> > fmast@...








> Richard W. Burry, Ph.D.
> Department of Neuroscience, College of Medicine
> Campus Microscopy and Imaging Facility, Director
> The Ohio State University
> Associate Editor, Journal of Histochemistry and Cytochemistry
> 277 Biomedical Research Tower
> 460 West Twelfth Avenue
> Columbus, Ohio 43210
> Voice 614.292.2814  Cell 614.638.3345  Fax 614.247.8849



> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM



> Spam
> Not spam
> Forget previous vote


> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM






No virus found in this incoming message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM


No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.552 / Virus Database: 270.10.20/1944 - Release Date: 10/02/2009 5:44 PM