To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocal community, Does anyone know of where I could buy discontinued or used (in very good condition) Zeiss Ultrafluor objectives, 40x or higher? It seems their higher magnifications are discontinued. Alternatively, if anybody knows other objectives with high transmission ratio down to 260 nm it would be greatly appreciated. Other deep UV threads on this list mostly discuss light sources and multiphoton excitation approaches - I did not see much discussion on sourcing UV imaging optics but apologies if I missed an existing, relevant thread. High NA reflective objectives are also an option, but it would seem that accessible off the shelf models are aberration-limited (for example thorlabs LMM-40X-UVV-160 includes a zemax model showing a ray spot diagram limited to 1.2 um resolution). I have not seen models of other similar objectives, but they are for the most part also two-mirror Schwarzschild layout and so also limited to two correcting surfaces. It occurs to me that the lithography community would be a great source of high resolution reflecting optics, but I am much less familiar with that field. Any help would be greatly appreciated. Thanks! Paul -- Paul Lebel, PhD Senior R&D Engineer Chan Zuckerberg Biohub 499 Illinois Street, San Francisco |
|
Tobias Baskin |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Paul, I have had good luck with this kind of thing (not exact thing) from Vermont Optechs. http://www.scopeshop.com. No finanical ties, just a happy customer. Good luck, Tobias On 10/4/17 12:01 PM, Paul Lebel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocal community, > Does anyone know of where I could buy discontinued or used (in very good > condition) Zeiss Ultrafluor objectives, 40x or higher? It seems their > higher magnifications are discontinued. Alternatively, if anybody knows > other objectives with high transmission ratio down to 260 nm it would be > greatly appreciated. Other deep UV threads on this list mostly discuss > light sources and multiphoton excitation approaches - I did not see much > discussion on sourcing UV imaging optics but apologies if I missed an > existing, relevant thread. > > High NA reflective objectives are also an option, but it would seem that > accessible off the shelf models are aberration-limited (for example > thorlabs LMM-40X-UVV-160 includes a zemax model showing a ray spot diagram > limited to 1.2 um resolution). I have not seen models of other similar > objectives, but they are for the most part also two-mirror Schwarzschild > layout and so also limited to two correcting surfaces. It occurs to me that > the lithography community would be a great source of high resolution > reflecting optics, but I am much less familiar with that field. > > Any help would be greatly appreciated. > Thanks! > > Paul > > -- __ ___ ^ ___ ___ Tobias I. Baskin / \ / / \ / \ Professor / / / / \ \ \ Biology Department / __/ /__ /___ \ \ \__ University of Mass. / / / \ \ \ 611 N. Pleasant St. / / / \ \ \ Amherst, Massachusetts / /___ / \ \___/ \_____ USA 01003 413-545-1533 www.bio.umass.edu/biology/baskin BLOG: blogs.umass.edu/baskin/ |
|
|
In reply to this post by Paul Lebel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** There are couple used Zeiss Ultrafluar on eBay 80x/0.95 and 120x/0.95 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Lebel Sent: Wednesday, October 04, 2017 12:02 PM To: [hidden email] Subject: Zeiss Ultrafluor objectives ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocal community, Does anyone know of where I could buy discontinued or used (in very good condition) Zeiss Ultrafluor objectives, 40x or higher? It seems their higher magnifications are discontinued. Alternatively, if anybody knows other objectives with high transmission ratio down to 260 nm it would be greatly appreciated. Other deep UV threads on this list mostly discuss light sources and multiphoton excitation approaches - I did not see much discussion on sourcing UV imaging optics but apologies if I missed an existing, relevant thread. High NA reflective objectives are also an option, but it would seem that accessible off the shelf models are aberration-limited (for example thorlabs LMM-40X-UVV-160 includes a zemax model showing a ray spot diagram limited to 1.2 um resolution). I have not seen models of other similar objectives, but they are for the most part also two-mirror Schwarzschild layout and so also limited to two correcting surfaces. It occurs to me that the lithography community would be a great source of high resolution reflecting optics, but I am much less familiar with that field. Any help would be greatly appreciated. Thanks! Paul -- Paul Lebel, PhD Senior R&D Engineer Chan Zuckerberg Biohub 499 Illinois Street, San Francisco |
|
Benjamin Smith |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If a small spot size is what you are after, a quick web search popped up the following objectives from the lithography field: https://www.corning.com/media/worldwide/csm/documents/c3e0616ab18a4af8a02b05e18d7631d3.pdf An older variant of this objective was sited in the following paper: https://ac.els-cdn.com/016793179390017Y/1-s2.0-016793179390017Y-main.pdf?_tid=47276092-a993-11e7-898e-00000aab0f6c&acdnat=1507183748_3367eb4bd65a27afe5bd792fe7b42b5f There is another good paper here using the 198nm objective for metrology: http://proceedings.spiedigitallibrary.org/pdfaccess.ashx?url=/data/conferences/spiep/24914/894_1.pdf Cheers, Ben Smith On Wed, Oct 4, 2017 at 9:39 PM, Alex GT <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > There are couple used Zeiss Ultrafluar on eBay 80x/0.95 and 120x/0.95 > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Paul Lebel > Sent: Wednesday, October 04, 2017 12:02 PM > To: [hidden email] > Subject: Zeiss Ultrafluor objectives > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocal community, > Does anyone know of where I could buy discontinued or used (in very good > condition) Zeiss Ultrafluor objectives, 40x or higher? It seems their > higher magnifications are discontinued. Alternatively, if anybody knows > other objectives with high transmission ratio down to 260 nm it would be > greatly appreciated. Other deep UV threads on this list mostly discuss > light sources and multiphoton excitation approaches - I did not see much > discussion on sourcing UV imaging optics but apologies if I missed an > existing, relevant thread. > > High NA reflective objectives are also an option, but it would seem that > accessible off the shelf models are aberration-limited (for example > thorlabs LMM-40X-UVV-160 includes a zemax model showing a ray spot diagram > limited to 1.2 um resolution). I have not seen models of other similar > objectives, but they are for the most part also two-mirror Schwarzschild > layout and so also limited to two correcting surfaces. It occurs to me that > the lithography community would be a great source of high resolution > reflecting optics, but I am much less familiar with that field. > > Any help would be greatly appreciated. > Thanks! > > Paul > > > -- > Paul Lebel, PhD > Senior R&D Engineer > Chan Zuckerberg Biohub > 499 Illinois Street, San Francisco > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] http://vision.berkeley.edu/?page_id=5635 <http://vision.berkeley.edu/> |
|
Johannes Helm |
|
|
In reply to this post by Paul Lebel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Paul, please, note that the name of these objective series was - and still is, although, as far as I know, production has been discontinued - "Ultrafluar" with "a", NOT "Ultraflu_o_r". When using E-bay or other sources, as had been suggested in an earlier response by user "Alex GT", one should nevertheless check out both spellings, since the wrong name "Ultrafluor" has, specifically in the anglophone world, been used for decades. Also, one should note that these objectives, albeit outstanding in transmission from the limits of VUV and to the IR, do exhibit comparatively strong axial and lateral chromatic aberrations. Also, while they are anastigmatic, indeed, the curvature of their field of view is comparatively strong. The source of this information is partly Dr. Höcherl at Zeiss, at this time quite certainly retired since many years, who, in the late 80s and early 90s let me know a lot of details about these lenses, when my task had been to built a CLSM for Fura-2 measurements. I had been lucky at that time to have a 32x Ultrafluar on loan from Dr. Thorbjörn Caspersson together with a Bausch & Lomb couple of 32x catadioptric objective and condenser from Dr. Rudolf Rigler, both at the Karolinska Inst. in Solna, Sweden. Both gentlemen at this time are retired, Dr. Caspersson had already been in his early 90s at that time. I do not have the slightest idea who is in charge of these marvelous lenses at this time. Also, one should take great care when buying old style Ultrafluars, which still had been made for 160mm tube length microscopes. They will exhibit strange properties when used on infinity corrected microscopes. If one, however, gets lenses of the newer type, made for infinity corrected microscopes, one should also make sure to get the appropriate tube lens. While the standard tube lens for the Axio-series of microscopes was a single lens - very tricky and intelligently adapted to the Zeiss objectives to minimize aberrations -, the tube lens for the Ultrafluars was an achromatic doublet. Using the Ultrafluars on wrong tube lengths microscopes or together with an inappropriate tube lens may render them more or less useless. Last and possibly not least: Some of these lenses were Glycerin immersion lenses made for coverslips from SiO2, 200microns in thickness. While fused silica is optically isotropic, good transmission values for the deep UV can be attained for coverslips from crystalline quarts, but these exhibit polarization effects due to the birefringence of the crystalline material. There was also a 100x/0.95 Ultrafluar dry lens for cover slip thickness 0, i.e. one for reflected light microscopy on metals or minerals. I am personally not sure whether there ever had been any 80x or 120x lenses by Zeiss. If these should be offered on E-bay, I would, indeed, carefully check the source. Best wishes, Johannes On 2017-10-04 18:01, Paul Lebel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear confocal community, > Does anyone know of where I could buy discontinued or used (in very > good > condition) Zeiss Ultrafluor objectives, 40x or higher? It seems their > higher magnifications are discontinued. Alternatively, if anybody knows > other objectives with high transmission ratio down to 260 nm it would > be > greatly appreciated. Other deep UV threads on this list mostly discuss > light sources and multiphoton excitation approaches - I did not see > much > discussion on sourcing UV imaging optics but apologies if I missed an > existing, relevant thread. > > High NA reflective objectives are also an option, but it would seem > that > accessible off the shelf models are aberration-limited (for example > thorlabs LMM-40X-UVV-160 includes a zemax model showing a ray spot > diagram > limited to 1.2 um resolution). I have not seen models of other similar > objectives, but they are for the most part also two-mirror > Schwarzschild > layout and so also limited to two correcting surfaces. It occurs to me > that > the lithography community would be a great source of high resolution > reflecting optics, but I am much less familiar with that field. > > Any help would be greatly appreciated. > Thanks! > > Paul -- P. Johannes Helm Voice: (+47) 228 51159 (office) Fax: (+47) 228 51499 (office) |
|
|
In reply to this post by Paul Lebel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Johannes, Thank you for your detailed reply, and also for the spelling correction. Especially useful to me are your tips on residual aberrations and the information about tube lens types. If the aberrations are largely chromatic in nature, I should be able to minimize their effects as I will be using LED illumination with a relatively narrow bandwidth (I could also filter it further if needed). Also, I have the flexibility of an open breadboard platform so I can adapt to either older style 160 mm finite conjugate or to infinity-corrected with a tube lens, depending on what I am able to procure. In the infinity-corrected case, do you know if the standard Zeiss tube lenses also transmit down to 260 nm? It would seem like they would need to do so in order to ensure compatibility with the Ultrafluar series, but at the same time it sounds very restrictive on their design, given the limited choice of glasses at that wavelength. Finally, yes I plan to use 200um quartz coverslips with my samples and birefringence is duly noted. Thanks also for the many other replies with suggested sources for Ultrafluar and lithography objectives! Regards, Paul |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** From memory: Getting 360nm laser on LSM510 confocal required major upgrades to Zeiss microscopes $$. So standard Zeiss lens would not handle 260nm. Alex Tkachenko Lab Director Zorbius BioPharma -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Lebel Sent: Friday, October 06, 2017 1:20 PM To: [hidden email] Subject: Re: Zeiss Ultrafluor objectives ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Johannes, Thank you for your detailed reply, and also for the spelling correction. Especially useful to me are your tips on residual aberrations and the information about tube lens types. If the aberrations are largely chromatic in nature, I should be able to minimize their effects as I will be using LED illumination with a relatively narrow bandwidth (I could also filter it further if needed). Also, I have the flexibility of an open breadboard platform so I can adapt to either older style 160 mm finite conjugate or to infinity-corrected with a tube lens, depending on what I am able to procure. In the infinity-corrected case, do you know if the standard Zeiss tube lenses also transmit down to 260 nm? It would seem like they would need to do so in order to ensure compatibility with the Ultrafluar series, but at the same time it sounds very restrictive on their design, given the limited choice of glasses at that wavelength. Finally, yes I plan to use 200um quartz coverslips with my samples and birefringence is duly noted. Thanks also for the many other replies with suggested sources for Ultrafluar and lithography objectives! Regards, Paul |
|
Johannes Helm |
|
|
In reply to this post by Paul Lebel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** On 2017-10-06 19:20, Paul Lebel wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear Johannes, > Thank you for your detailed reply, and also for the spelling > correction. Especially useful to me are your tips on residual > aberrations and the information about tube lens types. If the > aberrations are largely chromatic in nature, I should be able to > minimize their effects as I will be using LED illumination with a > relatively narrow bandwidth (I could also filter it further if > needed). Also, I have the flexibility of an open breadboard platform > so I can adapt to either older style 160 mm finite conjugate or to > infinity-corrected with a tube lens, depending on what I am able to > procure. In the infinity-corrected case, do you know if the standard > Zeiss tube lenses also transmit down to 260 nm? Hi, again, Paul, answer, unfortunately: No! As you correctly write, few refractive materials are available for designing optics to be used in the deep UV. While there might be more, today, than had been available during the 80s and 90s, at that time they had to live with a couple of fluorites and calcites. Also, if I remember correctly, neither the mechanical properties of the available materials, which had a tendency to easily crack when being formed, nor the tightness of the tolerances to be kept during lens fabrication left the craftsmen specifically happy, to place it mildly. Seemingly, designing was far less of a problem than building. The standard tube lens of the old Axio series - I do, and rather unfortunately so, have no clue about Zeiss' more and most modern optics - was good, as far as I remember, to somewhere between 380nm and 370nm, at which the limit for imaging purposes primarily was not set by the drop of the rate of transmission but rather by aberrations. As you quite certainly know, in most materials the refractive index changes per nm get larger and larger the more you approach the absorption limit. A plethora of interesting data is available from the Schott Glass Catalogue (and possibly other and similar sources). Best wishes and have a good success in your research! Johannes > It would seem like > they would need to do so in order to ensure compatibility with the > Ultrafluar series, but at the same time it sounds very restrictive on > their design, given the limited choice of glasses at that wavelength. > Finally, yes I plan to use 200um quartz coverslips with my samples and > birefringence is duly noted. > > Thanks also for the many other replies with suggested sources for > Ultrafluar and lithography objectives! > Regards, > > Paul -- P. Johannes Helm Voice: (+47) 228 51159 (office) Fax: (+47) 228 51499 (office) |
|
George McNamara |
|
|
In reply to this post by Alex GT
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Not a major upgrade if purchased up front. From my memory: I managed a Zeiss LSM510 (4 PMT) confocal microscope in Miami with a UV Argon ion 351 nm & 364 nm laser line. Yes, the laser was finicky (and we eventually retired it). My GUESStimates (and these are only gueses) is that Zeiss sold over 2000 LSM510s and probably 10% (i.e. 200) had this 351/364 nm laser (maybe more). On 10/6/2017 1:27 PM, Alex GT wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > From memory: Getting 360nm laser on LSM510 confocal required major upgrades to Zeiss microscopes $$. So standard Zeiss lens would not handle 260nm. > > Alex Tkachenko > Lab Director > Zorbius BioPharma > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Lebel > Sent: Friday, October 06, 2017 1:20 PM > To: [hidden email] > Subject: Re: Zeiss Ultrafluor objectives > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Johannes, > Thank you for your detailed reply, and also for the spelling correction. Especially useful to me are your tips on residual aberrations and the information about tube lens types. If the aberrations are largely chromatic in nature, I should be able to minimize their effects as I will be using LED illumination with a relatively narrow bandwidth (I could also filter it further if needed). Also, I have the flexibility of an open breadboard platform so I can adapt to either older style 160 mm finite conjugate or to infinity-corrected with a tube lens, depending on what I am able to procure. In the infinity-corrected case, do you know if the standard Zeiss tube lenses also transmit down to 260 nm? It would seem like they would need to do so in order to ensure compatibility with the Ultrafluar series, but at the same time it sounds very restrictive on their design, given the limited choice of glasses at that wavelength. Finally, yes I plan to use 200um quartz coverslips with my samples and birefringence is duly noted. > > Thanks also for the many other replies with suggested sources for Ultrafluar and lithography objectives! > Regards, > > Paul -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |