To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi , We have been working with ICG dyes to track blood flow in vivo using a 806nm laser light, however our new equipment does not work very well with such long wavelength, I am wondering if anyone knows about an alternative dye which binds with plasma protein but using a shorter excitation wavelength ( like in the upper visible region 660-735nm range ), thanks in advance. Alex |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've had some success using IRDye 680LT for confocal imaging. It has a fairly broad absorption spectrum and I've used it with a 640nm diode laser with some success. It is quite photostable as well, but you will have to see if they can get it conjugated for what you want. Craig Brideau On Thu, Mar 27, 2014 at 5:17 PM, Ko, Alex <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi , > > We have been working with ICG dyes to track blood flow in vivo using a > 806nm laser light, however our new equipment does not work very well with > such long wavelength, I am wondering if anyone knows about an alternative > dye which binds with plasma protein but using a shorter excitation > wavelength ( like in the upper visible region 660-735nm range ), thanks in > advance. > > Alex |
![]() ![]() |
Armstrong, Brian |
![]() |
In reply to this post by Ko, Alex
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, you could try dextran conjugates. Molecular Probes has dextrans in many different spectra. Of course dextran is a sugar (polysaccharide) and does not specifically bind to plasma. Cheers, Brian Armstrong PhD Director, Light Microscopy Core x62872 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ko, Alex Sent: Thursday, March 27, 2014 4:18 PM To: [hidden email] Subject: alternative dye for ICG using shorter excitation wavelength ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi , We have been working with ICG dyes to track blood flow in vivo using a 806nm laser light, however our new equipment does not work very well with such long wavelength, I am wondering if anyone knows about an alternative dye which binds with plasma protein but using a shorter excitation wavelength ( like in the upper visible region 660-735nm range ), thanks in advance. Alex --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |
![]() ![]() |
Rainer Kohler-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, We use Angiosense680 or Angiospark 680 but it is expensive. Rainer On Mar 27, 2014, at 7:42 PM, Armstrong, Brian <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, you could try dextran conjugates. Molecular Probes has dextrans in many different spectra. Of course dextran is a sugar (polysaccharide) and does not specifically bind to plasma. Cheers, Brian Armstrong PhD Director, Light Microscopy Core x62872 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ko, Alex Sent: Thursday, March 27, 2014 4:18 PM To: [hidden email] Subject: alternative dye for ICG using shorter excitation wavelength ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi , We have been working with ICG dyes to track blood flow in vivo using a 806nm laser light, however our new equipment does not work very well with such long wavelength, I am wondering if anyone knows about an alternative dye which binds with plasma protein but using a shorter excitation wavelength ( like in the upper visible region 660-735nm range ), thanks in advance. Alex --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- Rainer Kohler, Ph.D Microscopist Center for Systems Biology Massachusetts General Hospital Richard B. Simches Research Center 185 Cambridge Street Suite 5.210 Boston, MA 02114 Phone.: 617 643 6391 Cell phone: 978 578 5057 The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by Armstrong, Brian
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** http://www.perkinelmer.com/Catalog/Category/ID/Vascular AngioSense 680 etc (Alexa Fluor 680 on something) see also http://www.perkinelmer.com/Catalog/Category/ID/Fluorescence%20Agents%20Portfolio for lighting up more body parts. These are from VisEn Medical, which was bought by Caliper Life Sciences which was bought by PerkinElmer ... which we'll have to see if it gets bought by Thermo Fisher someday. At least they shouldn't need to divest their NIRF line since it is sourced from T.F. (and then GE Healthcare will but T.F. and Neutron Jack 2.0 will split out the loser business units). Enjoy, George On 3/27/2014 6:42 PM, Armstrong, Brian wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi, you could try dextran conjugates. Molecular Probes has dextrans in many different spectra. Of course dextran is a sugar (polysaccharide) and does not specifically bind to plasma. > Cheers, > > Brian Armstrong PhD > Director, Light Microscopy Core > x62872 > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ko, Alex > Sent: Thursday, March 27, 2014 4:18 PM > To: [hidden email] > Subject: alternative dye for ICG using shorter excitation wavelength > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi , > > We have been working with ICG dyes to track blood flow in vivo using a 806nm laser light, however our new equipment does not work very well with such long wavelength, I am wondering if anyone knows about an alternative dye which binds with plasma protein but using a shorter excitation wavelength ( like in the upper visible region 660-735nm range ), thanks in advance. > > Alex > > > --------------------------------------------------------------------- > *SECURITY/CONFIDENTIALITY WARNING: > This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) > --------------------------------------------------------------------- > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/26/ |