To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all We have little experience with refurbishing argon lasers. We only have 1 refurbished argon and this was done only a couple of months ago. Now that we are in need of refurbishing one again, our local vendor claims that refurbished argon lasers last a lot less time than new ones and often stop working after one year. I have search the list archives without finding any comparison of the lifetime of refurbished vs new argon lasers. What is your experience with refurbished argon lasers? Do they last less than new argons? thanks Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7C, Room 7362/7840 14157 Huddinge, Sweden mobile: +46 (0) 73 733 5008 LCI website<http://ki.se/en/bionut/welcome-to-the-lci-facility> |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The things that usually go on an Argon are the electrodes in the gas tube or the high-voltage power supply. When you get an argon refurbished the gas tube is replaced, so you get fresh electrodes, and a fresh batch of gas at the correct pressure. Assuming the cavity mirrors are relatively clean then the new tube basically makes it like a new laser. The exception is if you are having problems with the power supply, although the repair company can also deal with issues there by replacing capacitors, etc. Craig On Tue, Feb 6, 2018 at 4:21 AM, Sylvie Le Guyader <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all > > We have little experience with refurbishing argon lasers. We only have 1 > refurbished argon and this was done only a couple of months ago. > Now that we are in need of refurbishing one again, our local vendor claims > that refurbished argon lasers last a lot less time than new ones and often > stop working after one year. > I have search the list archives without finding any comparison of the > lifetime of refurbished vs new argon lasers. > > What is your experience with refurbished argon lasers? Do they last less > than new argons? > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7C, > Room 7362/7840 > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-facility> > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear microscopists, I am looking for a LSM510/710 sscanhead with optics/filters/detectors supporting at least 405/488/543/633 laser lines and two PMT detectors. A working condition is strongly preferred, as well as accompanying acquisition electronics module though non working condition also may be considered for cheap price - maybe it would be possible to combine parts from several into single working scanhead. Thanks, Arvydas ********************** Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core SUNY Upstate Medical University Neuroscience & Physiology Dept Syracuse, NY |
|
Arvydas Matiukas |
|
|
In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Microscopists, Please share you thoughts/experience what would be an inexpensive way to upgrade 510 META NLO on the upright Axioskop 2 FS mot stand. i am aiming at the most sensitive available detectors( Idealy 2x GaAsP for two colors). I would prefer keep the 2p Chameleon laser and the upright stand. By inexpensive plan I assume a fraction of the Thorlabs 2p scope price ($100k). Fitting into $20k budget would be perfect. Initially I was thinking to reconnect the priginal LSM510 META NLO scanhead to tothe upright stand, however, some people working with 2p setups pointed to me that the sensitivity of such system won't be good enough for the most deep tissue imaging projects. Therefore I am wondering if it is possible to significantly improve sensitivity just by replacing old PMT/NDD detectors, or optics/beampath still be an issue . Thanks, Arvydas *************************** Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core SUNY Upstate Medical University Neuroscience & Physiology Dept Syracuse, NY >>> Sylvie Le Guyader <[hidden email]> 2/6/2018 6:21 AM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all We have little experience with refurbishing argon lasers. We only have 1 refurbished argon and this was done only a couple of months ago. Now that we are in need of refurbishing one again, our local vendor claims that refurbished argon lasers last a lot less time than new ones and often stop working after one year. I have search the list archives without finding any comparison of the lifetime of refurbished vs new argon lasers. What is your experience with refurbished argon lasers? Do they last less than new argons? thanks Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7C, Room 7362/7840 14157 Huddinge, Sweden mobile: +46 (0) 73 733 5008 LCI website<http://ki.se/en/bionut/welcome-to-the-lci-facility> |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Arvydas, I suggest investing in GPU deconvolution, for example, www.microvolution.com or SVI Huygens (www.svi.nl) or AutoQuant. All three commercial software give comparable results (I trialed each summer 2017 on an NVidia GTX 1060 on my office PC ... and then another division dropped in an Leica SP8 + HyVolution2 [SVI] instrument to keep me busy). See the example image pair on top of page 2 (ignoring the cover page) of McTips 2017 download https://works.bepress.com/gmcnamara/81 I am still very interested in buying a GPU deconvolution license and having "drop&drag" instant gratification on our campus intranet ... ideally everyone who publishes would acknowledge our image core, parent P30 center grant and any S10 grants (and other funding sources) to fund instruments used in the research. With respect to S10 grants: my thanks to the study section that awarded our image core's proposal a score in the funding range (I started work a few days before it was submitted, so all the credit goes to Brian, Olga, Mark, Bin, John Gibas, our users, and Jason Brenner of Olympus [John is now also at Olympus]). Hopefully the new instrument will be installed and operational before the Microscopy Society of America (MSA) meeting in August here in Baltimore and we can have meeting attendees over. https://www.microscopy.org/MandM/2018 George p.s. disclosure: a very nice dataset that I 'clicked' is on the gallery at https://www.microvolution.com/gallery (no, Marc Bruce did not buy me a new Tesla, or even send me a matchbox Tesla, nor any other 'direct deposit'). I also recommend the TOP1 single molecule RNA FISH dataset (likely Stellaris RNA FISH probe set) acquired by Dane Maxfield. On 2/15/2018 2:06 PM, Arvydas Matiukas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello Microscopists, > > Please share you thoughts/experience what would be an inexpensive way > to upgrade > 510 META NLO on the upright Axioskop 2 FS mot stand. i am aiming at the > most sensitive > available detectors( Idealy 2x GaAsP for two colors). I would prefer > keep the 2p Chameleon laser > and the upright stand. By inexpensive plan I assume a fraction of the > Thorlabs 2p scope price ($100k). > Fitting into $20k budget would be perfect. > > Initially I was thinking to reconnect the priginal LSM510 META NLO > scanhead to tothe upright stand, > however, some people working with 2p setups pointed to me that the > sensitivity of such system won't > be good enough for the most deep tissue imaging projects. Therefore I > am wondering if it is possible > to significantly improve sensitivity just by replacing old PMT/NDD > detectors, or optics/beampath still be > an issue . > > Thanks, > Arvydas > *************************** > > > > > > > Arvydas Matiukas, Ph.D. > Director of Confocal&Two-Photon Core > SUNY Upstate Medical University > Neuroscience & Physiology Dept > > Syracuse, NY > >>> Sylvie Le Guyader <[hidden email]> 2/6/2018 6:21 AM >>> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear all > > We have little experience with refurbishing argon lasers. We only have > 1 refurbished argon and this was done only a couple of months ago. > Now that we are in need of refurbishing one again, our local vendor > claims that refurbished argon lasers last a lot less time than new ones > and often stop working after one year. > I have search the list archives without finding any comparison of the > lifetime of refurbished vs new argon lasers. > > What is your experience with refurbished argon lasers? Do they last > less than new argons? > > thanks > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader, PhD > Live Cell Imaging Facility Manager > Karolinska Institutet- Bionut Dpt > Hälsovägen 7C, > Room 7362/7840 > 14157 Huddinge, Sweden > mobile: +46 (0) 73 733 5008 > LCI website<http://ki.se/en/bionut/welcome-to-the-lci-facility> -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
|
Reece, Jeff (NIH/NIDDK) [E] |
|
|
In reply to this post by Arvydas Matiukas
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Arvydas, Try LSM Tech: http://lsmtech.com/computer-upgrade/ Kind Regards, Jeff -----Original Message----- From: Arvydas Matiukas [mailto:[hidden email]] Sent: Thursday, February 15, 2018 2:06 PM To: [hidden email] Subject: Re: upgrade of 2p system on Axioskop 2 FS mot stand ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello Microscopists, Please share you thoughts/experience what would be an inexpensive way to upgrade 510 META NLO on the upright Axioskop 2 FS mot stand. i am aiming at the most sensitive available detectors( Idealy 2x GaAsP for two colors). I would prefer keep the 2p Chameleon laser and the upright stand. By inexpensive plan I assume a fraction of the Thorlabs 2p scope price ($100k). Fitting into $20k budget would be perfect. Initially I was thinking to reconnect the priginal LSM510 META NLO scanhead to tothe upright stand, however, some people working with 2p setups pointed to me that the sensitivity of such system won't be good enough for the most deep tissue imaging projects. Therefore I am wondering if it is possible to significantly improve sensitivity just by replacing old PMT/NDD detectors, or optics/beampath still be an issue . Thanks, Arvydas *************************** Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core SUNY Upstate Medical University Neuroscience & Physiology Dept Syracuse, NY >>> Sylvie Le Guyader <[hidden email]> 2/6/2018 6:21 AM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all We have little experience with refurbishing argon lasers. We only have 1 refurbished argon and this was done only a couple of months ago. Now that we are in need of refurbishing one again, our local vendor claims that refurbished argon lasers last a lot less time than new ones and often stop working after one year. I have search the list archives without finding any comparison of the lifetime of refurbished vs new argon lasers. What is your experience with refurbished argon lasers? Do they last less than new argons? thanks Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7C, Room 7362/7840 14157 Huddinge, Sweden mobile: +46 (0) 73 733 5008 LCI website<http://ki.se/en/bionut/welcome-to-the-lci-facility> |