![]() ![]() |
Cromey, Douglas W - (dcromey) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/micro Home of: "Microscopy and Imaging Resources on the WWW" |
![]() ![]() |
Kilgore, Jason-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Autofluorescence can be especially bad for paraffin-embedded sections. I'm not sure liver is particularly worse, though, compared to many tissues. I've had very good luck with sodium borohydride washing to reduce autofluorescence, particularly in the green and red range. Another thing I've found to reduce it is to use Histoclear or other limonine solvents for deparaffinization, instead of xylenes. Another thing to try is to amplify your label, to help overcome the background, using biotin/streptavidin, Tyramide signal amplification, or Qdots. Hope that helps. Cheers, Jason -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cromey, Douglas W - (dcromey) Sent: Tuesday, June 04, 2013 10:24 AM To: [hidden email] Subject: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/micro Home of: "Microscopy and Imaging Resources on the WWW" |
![]() ![]() |
Mark A. Sanders |
![]() |
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Doug, The liver autofluorescence may be due to lipofuscin. Try treating with Sudan Black. Sudan Black binds lipophilic compartments and, being black, absorbs the excitation energy and quenches lipofuscin. Here is a paper: Schnell SA, Staines WA, Wessendorf MW. Reduction of lipofuscin-like autofluorescence in fluorescently labeled tissue. J Histochem Cytochem. 1999 47(6):719-30. Cheers, Mark **************************************************** Mark A. Sanders University of Minnesota Program Director Twin Cities Campus University Imaging Centers St. Paul office ph: 612-624-3454 Mpls office ph: 612-626-3645 fax: 612-624-1799 www.uic.umn.edu On Jun 4, 2013, at 12:23 PM, "Cromey, Douglas W - (dcromey)" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? > > Doug > > ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ > Douglas W. Cromey, M.S. - Associate Scientific Investigator > Dept. of Cellular & Molecular Medicine, University of Arizona > 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA > > office: AHSC 4212 email: [hidden email] > voice: 520-626-2824 fax: 520-626-2097 > > http://swehsc.pharmacy.arizona.edu/micro > Home of: "Microscopy and Imaging Resources on the WWW" |
![]() ![]() |
Armen Khatchadourian |
![]() |
In reply to this post by Kilgore, Jason-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I had a similar problem with paraffin-embedded pancreatic sections. But I was getting much less autofluorescence in the far red region, compared to green and red, hence I started using AlexaFluor 647 -labeled secondary antibodies. I don't know if this would work for liver though. Best, Armen ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Kilgore, Jason [[hidden email]] Sent: Tuesday, June 04, 2013 5:23 PM To: [hidden email] Subject: Re: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Autofluorescence can be especially bad for paraffin-embedded sections. I'm not sure liver is particularly worse, though, compared to many tissues. I've had very good luck with sodium borohydride washing to reduce autofluorescence, particularly in the green and red range. Another thing I've found to reduce it is to use Histoclear or other limonine solvents for deparaffinization, instead of xylenes. Another thing to try is to amplify your label, to help overcome the background, using biotin/streptavidin, Tyramide signal amplification, or Qdots. Hope that helps. Cheers, Jason -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cromey, Douglas W - (dcromey) Sent: Tuesday, June 04, 2013 10:24 AM To: [hidden email] Subject: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/micro Home of: "Microscopy and Imaging Resources on the WWW" |
![]() ![]() |
Martin Wessendorf-2 |
![]() |
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Doug-- On 6/4/2013 12:23 PM, Cromey, Douglas W - (dcromey) wrote: > I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? Several questions: 0) Are you sure it's autofluorescence and not a problem with your staining protocol? Do you see the same autofluorescence in sections that haven't been stained? 1) What does the autofluorescence look like? Is it small, bright dots or a uniform high background? 2) Can you see it with the fluorescein filter? With the rhodamine filter? With UV excitation? 3) What color is the autofluorescence under a wide-band UV filter? The cure depends on the cause. If you see a uniform high background, it's probably due to fixation, and treating the sections with sodium borohydride (NaBH4) as Jason Kilgore suggested should help. See: http://www.ncbi.nlm.nih.gov/pubmed/9765122 If the autofluorescence looks like small, bright dots that don't photobleach, show up using all your filter sets and look orange using a wide-band UV filter, it's probably lipofuscin. That can be reduced by treating with Sudan Black, as Mark Sanders suggested, or with CuSO4, which is described in the same paper. CuSO4 is the better choice if you're running your tissue through xylene after staining--the Sudan Black will be eluted by lipophilic solvents. Good luck! Martin Wessendorf -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
![]() |
In reply to this post by Armen Khatchadourian
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** As per http://cammer.net/historical/aif/instructions/immunofluor/protocol01.htm II. REDUCE BACKGROUND Incubate with NaIO4 (21.4 mg/ml in PBS) 15 min. followed by PBS wash Alternative bleaching solutions can be fresh NaBH4 or glycine followed by PBS wash As per http://cammer.net/historical/aif/instructions/immunofluor/troubleshoot.htm Quench after fixation and before staining. Try: 1. Incubate with NaIO4 (21.4 mg/ml in PBS) 15 min.; PBS wash. 2. NaBH4 (10mg/ml) in PBS approx. 4 repeated washes 15 min. each; NaBH4 should be dry until immediately before use; PBS wash. [More.] 3. Incubate with glycine; PBS wash. Screen for autofluorescence before staining. _________________________________________ Michael Cammer, Assistant Research Scientist Skirball Institute of Biomolecular Medicine Lab: (212) 263-3208 Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Armen Khatchadourian [[hidden email]] Sent: Tuesday, June 04, 2013 5:42 PM To: [hidden email] Subject: Re: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I had a similar problem with paraffin-embedded pancreatic sections. But I was getting much less autofluorescence in the far red region, compared to green and red, hence I started using AlexaFluor 647 -labeled secondary antibodies. I don't know if this would work for liver though. Best, Armen ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Kilgore, Jason [[hidden email]] Sent: Tuesday, June 04, 2013 5:23 PM To: [hidden email] Subject: Re: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Autofluorescence can be especially bad for paraffin-embedded sections. I'm not sure liver is particularly worse, though, compared to many tissues. I've had very good luck with sodium borohydride washing to reduce autofluorescence, particularly in the green and red range. Another thing I've found to reduce it is to use Histoclear or other limonine solvents for deparaffinization, instead of xylenes. Another thing to try is to amplify your label, to help overcome the background, using biotin/streptavidin, Tyramide signal amplification, or Qdots. Hope that helps. Cheers, Jason -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cromey, Douglas W - (dcromey) Sent: Tuesday, June 04, 2013 10:24 AM To: [hidden email] Subject: autofluorescence in liver tissue ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? Doug ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ Douglas W. Cromey, M.S. - Associate Scientific Investigator Dept. of Cellular & Molecular Medicine, University of Arizona 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA office: AHSC 4212 email: [hidden email] voice: 520-626-2824 fax: 520-626-2097 http://swehsc.pharmacy.arizona.edu/micro Home of: "Microscopy and Imaging Resources on the WWW" |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by Cromey, Douglas W - (dcromey)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Doug, Tyramide signal amplification, with 11 washes after the primary, 11 washes after the 2ndAb-HRP (need to re-titer the antibodies compared to other methods ... and keep the tissue submerged at all times - my thanks to Richard Burry for the tip - see his Immunocytochemistry book). For killing the HRP between rounds (if multicolor TSA is needed) see peroxAbolish info in Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue using laser scanning cytometry. <http://www.ncbi.nlm.nih.gov/pubmed/21929847> *Takahashi* H, Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker S, Burke GW, Tzakis AG, *Ichii* H. Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 16. PMID: 21929847 You are not required to use the kit (Mol Probes or PerkinElmer sell TSA kits) HRP - or other components (except the fluorescent tyramide of course). I still have not tried it (not needed so far), but am intrigued by the HRP80 at http://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html If anyone has experience - either good or bad - with HRP80 or HRP50 or HRP20 - please let us know here on the listserv or tell me privately. George p.s. with respect to HRP80 * TSA ... positive specimens might need a warning: "do not look through the eyepiece with remaining eye" (yes, that's a joke - eyesight should recover after a few minutes). If you don't do enough washes, everything will be blindingly bright. Also, liver may have endogenous biotin sticking out, so need to make sure to block well (and always do the controls) when using streptavidin conjugates like HRP80. On 6/4/2013 12:23 PM, Cromey, Douglas W - (dcromey) wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I am working with a lab looking at IF stained sections of formalin-fixed, paraffin embedded pieces of mouse or rat liver. If there was any actual staining in their first try on our confocal, it was largely washed out by the lovely and strong autofluorescence. Because their animal experiment takes 8 weeks, it's not trivial (or inexpensive) to get new tissue, so they have been using existing blocks. No one in their lab had the foresight to put up snap-frozen tissues for cryosectioning. I have them looking at this document: http://www.uhnres.utoronto.ca/facilities/wcif/PDF/Autofluorescence.pdf but I'm wondering if liver is always a problem with autofluorescence, or is there maybe something in this lab's sample prep protocol that's making it worse. Any ideas? > > Doug > > ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^ > Douglas W. Cromey, M.S. - Associate Scientific Investigator > Dept. of Cellular& Molecular Medicine, University of Arizona > 1501 N. Campbell Ave, Tucson, AZ 85724-5044 USA > > office: AHSC 4212 email: [hidden email] > voice: 520-626-2824 fax: 520-626-2097 > > http://swehsc.pharmacy.arizona.edu/micro > Home of: "Microscopy and Imaging Resources on the WWW" > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 |
![]() ![]() |
Csúcs Gábor-2 |
![]() |
In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, A potential new user would have interest in using the VT-Hawk confocal system from Visitech. Is there anyone who has/had experience with the system? On paper the specifications look very nice, but I'd really appreciate your comments (also privately if required). Thanks Gabor Light Microscopy and Screening Centre ETH Zurich, HPM G41 Schafmattstrasse 18 CH-8093, Zurich Switzerland Phone: +41 44 633 6221 Mobile: +41 79 7582158 |
Free forum by Nabble | Edit this page |