To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, we are using an inverted microscope and need to do long term transmission imaging in a 24-well dish. the whole central part of microscope is surrounded by a box keeping it at 37 degrees and providing CO2. so far, we fail to avoid condensation at the plastik cover of wells. the even more annoying thing is that the condensation pattern changes from very fine to bigger droplets within ca. 10 hours. while the small droplets are actually not so bad, the bigger droplets cause severe problems for the transmission imaging (introducing some lensing effects). taking the lid off seems no option, because the wells may dry out over 10-24 hours. we tried putting some oil on the cover (from the inside of course), hoping that it would be more hydrophobic and thus prevent condensation, but unfortunately that only helped for a few hours... Anyone any ideas? Thanks already!!! Tischi (ALMF@EMBL Heidelberg) |
![]() ![]() |
Romin, Yevgeniy/Sloan Kettering Institute |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We use these mebranes that have a hydrophobic side to place on top of dishes instead of lids. It prevents drying out and has no condensation. I haven't seen these for 24-well plate, but I'm sure they have them. we go ours through our Zeiss applications specialist (no commercial interest) Yevgeniy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Tischer Sent: Tuesday, February 21, 2012 4:43 AM To: [hidden email] Subject: avoid condensation?? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, we are using an inverted microscope and need to do long term transmission imaging in a 24-well dish. the whole central part of microscope is surrounded by a box keeping it at 37 degrees and providing CO2. so far, we fail to avoid condensation at the plastik cover of wells. the even more annoying thing is that the condensation pattern changes from very fine to bigger droplets within ca. 10 hours. while the small droplets are actually not so bad, the bigger droplets cause severe problems for the transmission imaging (introducing some lensing effects). taking the lid off seems no option, because the wells may dry out over 10-24 hours. we tried putting some oil on the cover (from the inside of course), hoping that it would be more hydrophobic and thus prevent condensation, but unfortunately that only helped for a few hours... Anyone any ideas? Thanks already!!! Tischi (ALMF@EMBL Heidelberg) ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
![]() |
In reply to this post by Christian Tischer
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Tischi; I've used similar setups where I put a layer of light mineral oil over the top of the media in the wells. This prevents evaporation (I've gone up to 48 hours), and the pH seemed to be stable. Some cells may not tolerate this, but it's a simple thing to try. Kathy Spencer The Scripps Research Institute -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Tischer Sent: Tuesday, February 21, 2012 1:43 AM To: [hidden email] Subject: avoid condensation?? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, we are using an inverted microscope and need to do long term transmission imaging in a 24-well dish. the whole central part of microscope is surrounded by a box keeping it at 37 degrees and providing CO2. so far, we fail to avoid condensation at the plastik cover of wells. the even more annoying thing is that the condensation pattern changes from very fine to bigger droplets within ca. 10 hours. while the small droplets are actually not so bad, the bigger droplets cause severe problems for the transmission imaging (introducing some lensing effects). taking the lid off seems no option, because the wells may dry out over 10-24 hours. we tried putting some oil on the cover (from the inside of course), hoping that it would be more hydrophobic and thus prevent condensation, but unfortunately that only helped for a few hours... Anyone any ideas? Thanks already!!! Tischi (ALMF@EMBL Heidelberg) |
![]() |
In reply to this post by Christian Tischer
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We cover the medium with a thin layer of mineral oil, which has been saturated with gas; in our case CO2/O2/N2, at 37 degrees, and don't use the lid. Our mineral oil is from Sigma, other mineral oils may extract ingredients from the medium. When saturated, gas exchange across the oil is fine, as the absorption coefficients of CO2 and O2 are similar to those in water (Kubie, JBC, 1927). We perform imaging for up to five days in this manner, with no loss of medium and no condensation. Jim Denegre The Jackson Laboratory On 2/21/12 4:42 AM, "Christian Tischer" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello, > >we are using an inverted microscope and need to do long term transmission >imaging in a 24-well dish. the whole central part of microscope is >surrounded by >a box keeping it at 37 degrees and providing CO2. > >so far, we fail to avoid condensation at the plastik cover of wells. the >even >more annoying thing is that the condensation pattern changes from very >fine to >bigger droplets within ca. 10 hours. while the small droplets are >actually not so >bad, the bigger droplets cause severe problems for the transmission >imaging >(introducing some lensing effects). > >taking the lid off seems no option, because the wells may dry out over >10-24 >hours. > >we tried putting some oil on the cover (from the inside of course), >hoping that >it would be more hydrophobic and thus prevent condensation, but >unfortunately that only helped for a few hours... > >Anyone any ideas? > >Thanks already!!! > >Tischi (ALMF@EMBL Heidelberg) |
![]() ![]() |
Boswell, Carl A - (cboswell) |
![]() |
In reply to this post by Christian Tischer
Hi Tischi,
Have you considered humidifying the heating/gas chamber? There are some very simple methods to keep humidity high in a closed environment, including running the gas through water using an aquarium aerator. Cotton balls or crumpled paper towels in shallow dishes provide lots of moist surface area. Good luck, C Carl A. Boswell 520-742-6131 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Tischer Sent: Tuesday, February 21, 2012 2:43 AM To: [hidden email] Subject: avoid condensation?? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, we are using an inverted microscope and need to do long term transmission imaging in a 24-well dish. the whole central part of microscope is surrounded by a box keeping it at 37 degrees and providing CO2. so far, we fail to avoid condensation at the plastik cover of wells. the even more annoying thing is that the condensation pattern changes from very fine to bigger droplets within ca. 10 hours. while the small droplets are actually not so bad, the bigger droplets cause severe problems for the transmission imaging (introducing some lensing effects). taking the lid off seems no option, because the wells may dry out over 10-24 hours. we tried putting some oil on the cover (from the inside of course), hoping that it would be more hydrophobic and thus prevent condensation, but unfortunately that only helped for a few hours... Anyone any ideas? Thanks already!!! Tischi (ALMF@EMBL Heidelberg) |