To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We use a simple resistive heater on the top, similar to rear window defogger on automobiles. This works for days and avoids the need for oils. Cheers Rich Richard Cole Research Scientist V Director: Advanced Light Microscopy Core Unit Wadsworth Center Research Assistant Professor Dept. of Biomedical Sciences School of Public Health State University of New York P.O. Box 509 Albany N.Y. 12201-0509 518-474-7048 Phone 518-474-4430 Fax Email <mailto:[hidden email]> [hidden email] Website <http://www.wadsworth.org/cores/alm/index.htm> www.wadsworth.org/cores/alm/index.htm IMPORTANT NOTICE: This e-mail and any attachments may contain confidential or sensitive information which is, or may be, legally privileged or otherwise protected by law from further disclosure. It is intended only for the addressee. If you received this in error or from someone who was not authorized to send it to you, please do not distribute, copy or use it or any attachments. Please notify the sender immediately by reply e-mail and delete this from your system. Thank you for your cooperation. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > >We use these mebranes that have a hydrophobic side to place on top of dishes instead of lids. It prevents drying out and has no condensation. I haven't seen these for 24-well plate, but I'm sure they have them. we go ours through our Zeiss applications specialist (no commercial interest) > >Yevgeniy > Hi Yevgeniy, thanks for the suggestion! can you do transmission microscopy through those membranes? Tischi |
|
|
In reply to this post by Richard Cole
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Jim, wow. thanks! that sounds very interesting indeed! what do you mean by "saturating the oil with gas"; i mean what does that mean in practice; do you put the oil "into gas" before using it??? Tischi -------------- We cover the medium with a thin layer of mineral oil, which has been saturated with gas; in our case CO2/O2/N2, at 37 degrees, and don't use the lid. Our mineral oil is from Sigma, other mineral oils may extract ingredients from the medium. When saturated, gas exchange across the oil is fine, as the absorption coefficients of CO2 and O2 are similar to those in water (Kubie, JBC, 1927). We perform imaging for up to five days in this manner, with no loss of medium and no condensation. Jim Denegre The Jackson Laboratory |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We saturate by leaving dishes of oil in a gassed incubator for at lease on hour before use. On 2/22/12 10:53 AM, "Christian Tischer" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi Jim, > >wow. thanks! > >that sounds very interesting indeed! > >what do you mean by "saturating the oil with gas"; i mean what does that >mean in practice; do you put the oil "into gas" before using it??? > >Tischi > > > >-------------- > >We cover the medium with a thin layer of mineral oil, which has been >saturated with gas; in our case CO2/O2/N2, at 37 degrees, and don't use >the lid. Our mineral oil is from Sigma, other mineral oils may extract >ingredients from the medium. When saturated, gas exchange across the oil >is fine, as the absorption coefficients of CO2 and O2 are similar to those >in water (Kubie, JBC, 1927). We perform imaging for up to five days in >this manner, with no loss of medium and no condensation. > >Jim Denegre >The Jackson Laboratory |
| Free forum by Nabble | Edit this page |