Hi all,
        Does anyone know of a commerical source of fluorescent beads
that are around 25 microns in diameter? We are sorting large cells
and need something in that size range to calibrate the instrument.

        Many thanks,
                   Tobias
--
       _      ____          __   ____
      /  \   /          / \    /   \ \       Tobias I. Baskin
     /   /  /          /   \   \      \        Professor
    /_ /   __      /__ \   \       \__    Biology Department
   /      /          /       \   \       \        611 N. Pleasant St.
  /      /          /         \   \       \      University of
Massachusetts
/      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
www.bio.umass.edu/biology/baskin
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243
Tobias Baskin Tobias Baskin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: big fluorescent beads

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Gary (and anyone following this thread),
                        Thanks. As far I can
tell, introgen's beads top our at 15 µm. Well
that might be ok, but I was hoping to find 25 µm
(it is our average cell diameter). Sorry I didn't
make that clear (or perhaps there is a product I
missed on their web page amid all of the smaller
beads?).

        As ever
                Tobias

At 1:46 PM -0400 3/27/12, Gary G. Li wrote:

>Invitrogen?
>
>On Tue, Mar 27, 2012 at 1:25 PM, Tobias Baskin
><<mailto:[hidden email]>[hidden email]>
>wrote:
>
>*****
>To join, leave or search the confocal microscopy listserv, go to:
><http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>Hi all,
>        Does anyone know of a commerical source
>of fluorescent beads that are around 25 microns
>in diameter? We are sorting large cells and need
>something in that size range to calibrate the
>instrument.
>
>        Many thanks,
>                   Tobias
>--
>      _      ____          __   ____
>     /  \   /          / \    /   \ \       Tobias I. Baskin
>    /   /  /          /   \   \      \        Professor
>   /_ /   __      /__ \   \       \__    Biology Department
>  /      /          /       \   \       \        611 N. Pleasant St.
>  /      /          /         \   \       \
> University of Massachusetts
>/      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
><http://www.bio.umass.edu/biology/baskin>www.bio.umass.edu/biology/baskin
>Voice: <tel:413%20-%20545%20-%201533>413 - 545 -
>1533 Fax: <tel:413%20-%20545%20-%203243>413 -
>545 - 3243
abhijit debroy abhijit debroy
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: big fluorescent beads

In reply to this post by Tobias Baskin
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You may check spherotech. They appear to have particles up to 35 µm.
Catalog link below.

http://spherotech.com/five_up.htm

Thanks
Abhijit Deb Roy
Biomedical Sciences Program
University of Connecticut Health Centre, Farmington




On Tue, Mar 27, 2012 at 1:58 PM, Tobias Baskin <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/**wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
> *****
>
> Gary (and anyone following this thread),
>                        Thanks. As far I can tell, introgen's beads top our
> at 15 µm. Well that might be ok, but I was hoping to find 25 µm (it is our
> average cell diameter). Sorry I didn't make that clear (or perhaps there is
> a product I missed on their web page amid all of the smaller beads?).
>
>        As ever
>                Tobias
>
> At 1:46 PM -0400 3/27/12, Gary G. Li wrote:
>
>> Invitrogen?
>>
>>
>> On Tue, Mar 27, 2012 at 1:25 PM, Tobias Baskin <<mailto:
>> [hidden email]>**[hidden email] <[hidden email]>>
>> wrote:
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> <http://lists.umn.edu/cgi-bin/**wa?A0=confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
>> >http:**//lists.umn.edu/cgi-bin/wa?A0=**confocalmicroscopy<http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>
>>
>> *****
>>
>> Hi all,
>>       Does anyone know of a commerical source of fluorescent beads that
>> are around 25 microns in diameter? We are sorting large cells and need
>> something in that size range to calibrate the instrument.
>>
>>       Many thanks,
>>                  Tobias
>> --
>>     _      ____          __   ____
>>    /  \   /          / \    /   \ \       Tobias I. Baskin
>>   /   /  /          /   \   \      \        Professor
>>  /_ /   __      /__ \   \       \__    Biology Department
>>  /      /          /       \   \       \        611 N. Pleasant St.
>>  /      /          /         \   \       \ University of Massachusetts
>> /      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
>> <http://www.bio.umass.edu/**biology/baskin<http://www.bio.umass.edu/biology/baskin>
>> >www.bio.umass.**edu/biology/baskin<http://www.bio.umass.edu/biology/baskin>
>> Voice: <tel:413%20-%20545%20-%201533>413 - 545 - 1533 Fax:
>> <tel:413%20-%20545%20-%203243>413 - 545 - 3243
>>
>


--
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: big fluorescent beads

In reply to this post by Tobias Baskin
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Polysciences has fluorescent polystyrene in that size range:
http://www.polysciences.com/Catalog/Department/81/categoryId__338/

Kisker (Discover science) has 10micron fluorescent silica and 20 micron biotinylated that should be easy to label with fluorescent avidin.  
http://www.microparticles-nanospheres.com/micro_silicia.htm

Bang has 10-25 micron fluorescent
http://www.bangslabs.com/products/fluorescent_microspheres


On Mar 27, 2012, at 6:25 PM, Tobias Baskin wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
> Does anyone know of a commerical source of fluorescent beads
> that are around 25 microns in diameter? We are sorting large cells
> and need something in that size range to calibrate the instrument.
>
> Many thanks,
>   Tobias
> --
>       _      ____          __   ____
>      /  \   /          / \    /   \ \       Tobias I. Baskin
>     /   /  /          /   \   \      \        Professor
>    /_ /   __      /__ \   \       \__    Biology Department
>   /      /          /       \   \       \        611 N. Pleasant St.
>  /      /          /         \   \       \      University of
> Massachusetts
> /      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
> www.bio.umass.edu/biology/baskin
> Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243

Visiting Professor David Knecht
Beatson Institute for Cancer Research
University of Glasgow
Switchback Road, Bearsden
Glasgow Scotland G61 1BD
UK







Visiting Professor David Knecht
Beatson Institute for Cancer Research
University of Glasgow
Switchback Road, Bearsden
Glasgow Scotland G61 1BD
UK
Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: big fluorescent beads

In reply to this post by Tobias Baskin
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Check out http://bangslabs.com/ - they may have what you want. tom

Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Tobias Baskin
Sent: Tuesday, March 27, 2012 12:58 PM
To: [hidden email]
Subject: Re: big fluorescent beads

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Gary (and anyone following this thread),
                        Thanks. As far I can
tell, introgen's beads top our at 15 µm. Well that might be ok, but I was hoping to find 25 µm (it is our average cell diameter). Sorry I didn't make that clear (or perhaps there is a product I missed on their web page amid all of the smaller beads?).

        As ever
                Tobias

At 1:46 PM -0400 3/27/12, Gary G. Li wrote:

>Invitrogen?
>
>On Tue, Mar 27, 2012 at 1:25 PM, Tobias Baskin
><<mailto:[hidden email]>[hidden email]>
>wrote:
>
>*****
>To join, leave or search the confocal microscopy listserv, go to:
><http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>http://lists.umn
>.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>Hi all,
>        Does anyone know of a commerical source of fluorescent beads
>that are around 25 microns in diameter? We are sorting large cells and
>need something in that size range to calibrate the instrument.
>
>        Many thanks,
>                   Tobias
>--
>      _      ____          __   ____
>     /  \   /          / \    /   \ \       Tobias I. Baskin
>    /   /  /          /   \   \      \        Professor
>   /_ /   __      /__ \   \       \__    Biology Department
>  /      /          /       \   \       \        611 N. Pleasant St.
>  /      /          /         \   \       \
> University of Massachusetts
>/      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
><http://www.bio.umass.edu/biology/baskin>www.bio.umass.edu/biology/bask
>in
>Voice: <tel:413%20-%20545%20-%201533>413 - 545 -
>1533 Fax: <tel:413%20-%20545%20-%203243>413 -
>545 - 3243
Keith Morris Keith Morris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: big fluorescent beads

In reply to this post by Knecht, David
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

As David has said, do check out Polysciences:

http://www.polysciences.com/Catalog/40/categoryId__4

They used to manufacture beads to your specification if required, and I
don't think that has changed or that it was that expensive - but it's been
many years since I last did it.

Regards

Keith

---------------------------------------------------------------------------
Dr Keith J. Morris,
Molecular Cytogenetics and Microscopy Core,
Laboratory 00/069 and 00/070,
The Wellcome Trust Centre for Human Genetics,
Roosevelt Drive,
Oxford  OX3 7BN,
United Kingdom.

Telephone:  +44 (0)1865 287568
Email:  [hidden email]
Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of David Knecht
Sent: 27 March 2012 19:10
To: [hidden email]
Subject: Re: big fluorescent beads

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Polysciences has fluorescent polystyrene in that size range:
http://www.polysciences.com/Catalog/Department/81/categoryId__338/

Kisker (Discover science) has 10micron fluorescent silica and 20 micron
biotinylated that should be easy to label with fluorescent avidin.  
http://www.microparticles-nanospheres.com/micro_silicia.htm

Bang has 10-25 micron fluorescent
http://www.bangslabs.com/products/fluorescent_microspheres


On Mar 27, 2012, at 6:25 PM, Tobias Baskin wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
> Does anyone know of a commerical source of fluorescent beads
> that are around 25 microns in diameter? We are sorting large cells
> and need something in that size range to calibrate the instrument.
>
> Many thanks,
>   Tobias
> --
>       _      ____          __   ____
>      /  \   /          / \    /   \ \       Tobias I. Baskin
>     /   /  /          /   \   \      \        Professor
>    /_ /   __      /__ \   \       \__    Biology Department
>   /      /          /       \   \       \        611 N. Pleasant St.
>  /      /          /         \   \       \      University of
> Massachusetts
> /      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
> www.bio.umass.edu/biology/baskin
> Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243

Visiting Professor David Knecht
Beatson Institute for Cancer Research
University of Glasgow
Switchback Road, Bearsden
Glasgow Scotland G61 1BD
UK







Visiting Professor David Knecht
Beatson Institute for Cancer Research
University of Glasgow
Switchback Road, Bearsden
Glasgow Scotland G61 1BD
UK
John Oreopoulos John Oreopoulos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Small multi-color fluorescent beads (was "big fluorescent beads")

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Somewhat related to this last posting, can anyone recommend a source of small (50 nm diameter or less) multi-color (at least two distinct emission spectra) fluorescent microspheres? Commercial vendors may also contact me directly offline.  

Thank you.

John Oreopoulos
Research Assistant
Spectral Applied Research
Richmond Hill, Ontario
Canada
www.spectral.ca


On 2012-03-28, at 4:59 AM, Keith Morris wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> As David has said, do check out Polysciences:
>
> http://www.polysciences.com/Catalog/40/categoryId__4
>
> They used to manufacture beads to your specification if required, and I
> don't think that has changed or that it was that expensive - but it's been
> many years since I last did it.
>
> Regards
>
> Keith
>
> ---------------------------------------------------------------------------
> Dr Keith J. Morris,
> Molecular Cytogenetics and Microscopy Core,
> Laboratory 00/069 and 00/070,
> The Wellcome Trust Centre for Human Genetics,
> Roosevelt Drive,
> Oxford  OX3 7BN,
> United Kingdom.
>
> Telephone:  +44 (0)1865 287568
> Email:  [hidden email]
> Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of David Knecht
> Sent: 27 March 2012 19:10
> To: [hidden email]
> Subject: Re: big fluorescent beads
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Polysciences has fluorescent polystyrene in that size range:
> http://www.polysciences.com/Catalog/Department/81/categoryId__338/
>
> Kisker (Discover science) has 10micron fluorescent silica and 20 micron
> biotinylated that should be easy to label with fluorescent avidin.  
> http://www.microparticles-nanospheres.com/micro_silicia.htm
>
> Bang has 10-25 micron fluorescent
> http://www.bangslabs.com/products/fluorescent_microspheres
>
>
> On Mar 27, 2012, at 6:25 PM, Tobias Baskin wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi all,
>> Does anyone know of a commerical source of fluorescent beads
>> that are around 25 microns in diameter? We are sorting large cells
>> and need something in that size range to calibrate the instrument.
>>
>> Many thanks,
>>   Tobias
>> --
>>      _      ____          __   ____
>>     /  \   /          / \    /   \ \       Tobias I. Baskin
>>    /   /  /          /   \   \      \        Professor
>>   /_ /   __      /__ \   \       \__    Biology Department
>>  /      /          /       \   \       \        611 N. Pleasant St.
>> /      /          /         \   \       \      University of
>> Massachusetts
>> /      / ___   /           \   \__/  \ ____ Amherst, MA, 01003
>> www.bio.umass.edu/biology/baskin
>> Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243
>
> Visiting Professor David Knecht
> Beatson Institute for Cancer Research
> University of Glasgow
> Switchback Road, Bearsden
> Glasgow Scotland G61 1BD
> UK
>
>
>
>
>
>
>
> Visiting Professor David Knecht
> Beatson Institute for Cancer Research
> University of Glasgow
> Switchback Road, Bearsden
> Glasgow Scotland G61 1BD
> UK