		Hi all,
 I have some emission filters with dust that will not blow off, and one with an obvious fingerprint, all of which show up as blurry smutz in the image. I appreciate the need for caution when cleaning filters with soft coatings; e.g. no organic solvents, no touching, etc. But what about a simple soak in dilute soap and water, followed by dH2O rinse? Or, will a polymerizing cleaners, such as Photonic's First Contact work? Obviously there is a solvent in this to keep the material liquid until applied. Thanks, c Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 | 
|   | 
		Carl,
 Yes, you have to be very careful cleaning optics like this that have special coatings that affect their function. The only advice I have is this: A couple of times our lab has ordered some specialized dichroic mirrors and emission filters from Chroma and when they arrive in their box they come with a small orange slip of paper that talks about cleaning them. This slip says that if there a smudge of something that you really need to get remove that can't be taken off by a stream of air, just use a piece of lens paper and put a few drops of warm water on it. Then lightly drag this wet tissue across the optic's surface in one direction slowly. Repeat if necessary. As I recall, this topic of cleaning filters has come up on the server before about a year ago or less and several people had recommended the First Contact polymer. Try searching the archives on this for more information. (No commercial interest for Chroma or Photonic) John Oreopoulos On 7-Oct-09, at 1:42 PM, Carl Boswell wrote: > Hi all, > I have some emission filters with dust that will not blow off, and > one with an obvious fingerprint, all of which show up as blurry > smutz in the image. I appreciate the need for caution when cleaning > filters with soft coatings; e.g. no organic solvents, no touching, > etc. But what about a simple soak in dilute soap and water, > followed by dH2O rinse? Or, will a polymerizing cleaners, such as > Photonic's First Contact work? Obviously there is a solvent in > this to keep the material liquid until applied. > > Thanks, > > c > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 | 
|   | 
				In reply to this post by Boswell, Carl A - (cboswell)
			 
		I usually clean coated optics with a single swipe of methanol on a
 lens tissue. Take a lens tissue and fold it a few times, put a drop or two of methanol on it, and do a single swipe across the surface. If the fingerprint refuses to come off, refold the tissue you just used to expose a clean surface, add another drop of methanol and swipe again. If you are uncertain that your particular coatings are Methanol-compatible then test a tiny corner of the optic first! Also, make sure to use 'molecular grade' or ultra-high purity methanol, as standard stuff contains things that will contaminate your optic. Craig On Wed, Oct 7, 2009 at 11:42 AM, Carl Boswell <[hidden email]> wrote: > Hi all, > I have some emission filters with dust that will not blow off, and one with > an obvious fingerprint, all of which show up as blurry smutz in the image. I > appreciate the need for caution when cleaning filters with soft coatings; > e.g. no organic solvents, no touching, etc. But what about a simple soak in > dilute soap and water, followed by dH2O rinse? Or, will a polymerizing > cleaners, such as Photonic's First Contact work? Obviously there is a > solvent in this to keep the material liquid until applied. > > Thanks, > > c > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 > | 
|   | 
		I would agree with Craig on this.  Whether it is a filter or a dichroic
 mirror, "soft-coated" or "hard-coated" filter, the preferred method is first with compressed dry air and perhaps a soft lense cloth. If that fails to remove the smudge(s) a small amount of MeOH, EtOH, or even Acetone will not affect the coating if it is not repeated a multitude of times. Remember, "soft-coated" filters are not exposed coatings, so there is either just glass on the filter surface, or an AR coating. "Hard-coated" filters many times are exposed but are more resistant to solvent attack. At least by the ones mentioned here. Water cleaning can also damage filters and as water does not evaporate as quickly as the organics, if any gets behind the filter ring it will hang around longer. However, some grime does come off better in aqueous solvents, and it is OK to use a good breath of air on the filter surface and either a Qtip or wipe to clean it. A water dampened Qtip or cloth followed by a dry wipe should be fine as well. -Dan Dan Osborn Product Marketing Manager Omega Optical, Inc. Delta Campus Omega Drive Brattleboro, VT 05301 Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 Fax: 802-254-3937 Email: [hidden email] Web: www.omegafilters.com Celebrating 40 Years -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 1:53 PM To: [hidden email] Subject: Re: cleaning of filters I usually clean coated optics with a single swipe of methanol on a lens tissue. Take a lens tissue and fold it a few times, put a drop or two of methanol on it, and do a single swipe across the surface. If the fingerprint refuses to come off, refold the tissue you just used to expose a clean surface, add another drop of methanol and swipe again. If you are uncertain that your particular coatings are Methanol-compatible then test a tiny corner of the optic first! Also, make sure to use 'molecular grade' or ultra-high purity methanol, as standard stuff contains things that will contaminate your optic. Craig On Wed, Oct 7, 2009 at 11:42 AM, Carl Boswell <[hidden email]> wrote: > Hi all, > I have some emission filters with dust that will not blow off, and one > with an obvious fingerprint, all of which show up as blurry smutz in > the image. I appreciate the need for caution when cleaning filters > with soft coatings; e.g. no organic solvents, no touching, etc. But > what about a simple soak in dilute soap and water, followed by dH2O > rinse? Or, will a polymerizing cleaners, such as Photonic's First > Contact work? Obviously there is a solvent in this to keep the material liquid until applied. > > Thanks, > > c > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 > | 
|   | 
		Another problem with using water is if it is not ultra-pure you will
 leave residue behind. The same is true for using any of the alcohols; they need to be ultra high purity to avoid leaving behind gunk when they evaporate. Fingerprints are remarkably stubborn. It usually takes me two or three wipes with methanol to get them off. Dan's comment about using compressed clean air first is also very important. This takes the big particulate off the filter before you try to wipe it. That way you are not dragging 'gravel' over the surface of your optic when you wipe with lens tissue. Also, Kimwipes are not lens tissue. Use real lens tissue. Craig On Wed, Oct 7, 2009 at 12:20 PM, Dan Osborn <[hidden email]> wrote: > I would agree with Craig on this. Whether it is a filter or a dichroic > mirror, "soft-coated" or "hard-coated" filter, the preferred method is first > with compressed dry air and perhaps a soft lense cloth. If that fails to > remove the smudge(s) a small amount of MeOH, EtOH, or even Acetone will not > affect the coating if it is not repeated a multitude of times. Remember, > "soft-coated" filters are not exposed coatings, so there is either just > glass on the filter surface, or an AR coating. "Hard-coated" filters many > times are exposed but are more resistant to solvent attack. At least by the > ones mentioned here. > Water cleaning can also damage filters and as water does not evaporate as > quickly as the organics, if any gets behind the filter ring it will hang > around longer. > However, some grime does come off better in aqueous solvents, and it is OK > to use a good breath of air on the filter surface and either a Qtip or wipe > to clean it. A water dampened Qtip or cloth followed by a dry wipe should > be fine as well. > -Dan > > > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Craig Brideau > Sent: Wednesday, October 07, 2009 1:53 PM > To: [hidden email] > Subject: Re: cleaning of filters > > I usually clean coated optics with a single swipe of methanol on a lens > tissue. Take a lens tissue and fold it a few times, put a drop or two of > methanol on it, and do a single swipe across the surface. > If the fingerprint refuses to come off, refold the tissue you just used to > expose a clean surface, add another drop of methanol and swipe again. If > you are uncertain that your particular coatings are Methanol-compatible then > test a tiny corner of the optic first! > Also, make sure to use 'molecular grade' or ultra-high purity methanol, as > standard stuff contains things that will contaminate your optic. > > Craig > > > On Wed, Oct 7, 2009 at 11:42 AM, Carl Boswell <[hidden email]> > wrote: >> Hi all, >> I have some emission filters with dust that will not blow off, and one >> with an obvious fingerprint, all of which show up as blurry smutz in >> the image. I appreciate the need for caution when cleaning filters >> with soft coatings; e.g. no organic solvents, no touching, etc. But >> what about a simple soak in dilute soap and water, followed by dH2O >> rinse? Or, will a polymerizing cleaners, such as Photonic's First >> Contact work? Obviously there is a solvent in this to keep the material > liquid until applied. >> >> Thanks, >> >> c >> >> Carl A. Boswell, Ph.D. >> Molecular and Cellular Biology >> University of Arizona >> 520-954-7053 >> FAX 520-621-3709 >> > | 
|   | 
				In reply to this post by Craig Brideau
			 
		We also use MaOH, but never reuse a tissue. One tissue, one drop, one 
 wipe. If using your fingers to hold/fold the tissue, you need to be gloved to stop finger grease entering the tissue. The test on a corner as suggested is a good idea but so far I've never seen a problem with MeOH. Cheers > I usually clean coated optics with a single swipe of methanol on a > lens tissue. Take a lens tissue and fold it a few times, put a drop > or two of methanol on it, and do a single swipe across the surface. > If the fingerprint refuses to come off, refold the tissue you just > used to expose a clean surface, add another drop of methanol and swipe > again. If you are uncertain that your particular coatings are > Methanol-compatible then test a tiny corner of the optic first! > Also, make sure to use 'molecular grade' or ultra-high purity > methanol, as standard stuff contains things that will contaminate your > optic. > > Craig > > > On Wed, Oct 7, 2009 at 11:42 AM, Carl Boswell > <[hidden email]> wrote: > >> Hi all, >> I have some emission filters with dust that will not blow off, and one with >> an obvious fingerprint, all of which show up as blurry smutz in the image. I >> appreciate the need for caution when cleaning filters with soft coatings; >> e.g. no organic solvents, no touching, etc. But what about a simple soak in >> dilute soap and water, followed by dH2O rinse? Or, will a polymerizing >> cleaners, such as Photonic's First Contact work? Obviously there is a >> solvent in this to keep the material liquid until applied. >> >> Thanks, >> >> c >> >> Carl A. Boswell, Ph.D. >> Molecular and Cellular Biology >> University of Arizona >> 520-954-7053 >> FAX 520-621-3709 >> >> | 
|   | 
				In reply to this post by Dan Osborn
			 
		Dan Osborn wrote:
 > However, some grime does come off better in aqueous solvents, and it is OK > to use a good breath of air on the filter surface and either a Qtip or wipe > to clean it. A water dampened Qtip or cloth followed by a dry wipe should > be fine as well. I had always heard that Q-tips have starch in them as a binder, and were thus unsuitable for optics--that we should make our own with cotton wool and an applicator stick. Not so? Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] | 
|   | 
		I have unsubscribed twice following the original posting information. Does 
 anyone have an updated way to do this? --On Wednesday, October 07, 2009 2:41 PM -0500 Martin Wessendorf <[hidden email]> wrote: Dan Osborn wrote: > However, some grime does come off better in aqueous solvents, and it is OK > to use a good breath of air on the filter surface and either a Qtip or > wipe to clean it. A water dampened Qtip or cloth followed by a dry wipe > should be fine as well. I had always heard that Q-tips have starch in them as a binder, and were thus unsuitable for optics--that we should make our own with cotton wool and an applicator stick. Not so? Martin -- Patricia Masso-Welch, Ph.D. Dept. Biotechnical and Clinical Laboratory Sciences School of Medicine and Biomedical Sciences University of Buffalo 3435 Main Street Buffalo, NY 14214 Ph 716 829-5191 Fax 716 829-3601 [hidden email] | 
|   | 
				In reply to this post by Martin Wessendorf-2
			 
		The most common Q tips have both a binder to hold the cotton together
 and a glue that attaches the cotton to the stick. Either of these could dissolve in solvents and redeposit on optics. However, there is a glue-free, binder-free swab that you can special order, the 869-WC from Puritan: http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC I've used these with success on antireflection-coated singlets and doublets, although I haven't tested them on exposed filter coatings. One objection to these is that (lacking binder) they shed cotton fibers. In my experience those can be blown away with a stream of pressurized air. No financial interest. Deron Walters R&D Scientist, Physics Asylum Research > On Behalf Of Martin Wessendorf > > Dan Osborn wrote: > > > However, some grime does come off better in aqueous solvents, and it > > is OK to use a good breath of air on the filter surface and either a > > Qtip or wipe to clean it. A water dampened Qtip or cloth followed by > > a dry wipe should be fine as well. > > I had always heard that Q-tips have starch in them as a binder, and were > thus unsuitable for optics--that we should make our own with cotton wool > and an applicator stick. Not so? | 
|   | 
		Martin/ Deron,
 We use several of the Puritan products on protected coatings and mirrors: the same product as Deron listed, and 806 and 826 WC item numbers. The 869 was called out as the preferred product. And it true about the fibers, there is tendency for some to be left behind after cleaning, but usually a circular swabbing motion from the inside out can reduce this. There is also a fine polyester cleaning wipe used by some, but I do not have the catalog number on that. Best, Dan Dan Osborn Product Marketing Manager Omega Optical, Inc. Delta Campus Omega Drive Brattleboro, VT 05301 Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 Fax: 802-254-3937 Email: [hidden email] Web: www.omegafilters.com Celebrating 40 Years -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Deron Walters Sent: Wednesday, October 07, 2009 4:02 PM To: [hidden email] Subject: Re: cleaning of filters The most common Q tips have both a binder to hold the cotton together and a glue that attaches the cotton to the stick. Either of these could dissolve in solvents and redeposit on optics. However, there is a glue-free, binder-free swab that you can special order, the 869-WC from Puritan: http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC I've used these with success on antireflection-coated singlets and doublets, although I haven't tested them on exposed filter coatings. One objection to these is that (lacking binder) they shed cotton fibers. In my experience those can be blown away with a stream of pressurized air. No financial interest. Deron Walters R&D Scientist, Physics Asylum Research > On Behalf Of Martin Wessendorf > > Dan Osborn wrote: > > > However, some grime does come off better in aqueous solvents, and it > > is OK to use a good breath of air on the filter surface and either a > > Qtip or wipe to clean it. A water dampened Qtip or cloth followed by > > a dry wipe should be fine as well. > > I had always heard that Q-tips have starch in them as a binder, and were > thus unsuitable for optics--that we should make our own with cotton wool > and an applicator stick. Not so? | 
|    | Jerry (Gerald) Sedgewick | 
|   | 
				In reply to this post by Patricia Masso-Welch
			 
		Hi Patricia,
 Unsubscribe info is as follows: Send an email to [hidden email] and put the following into the body of the message: *signoff confocalmicroscopy* Patricia Masso-Welch wrote: > I have unsubscribed twice following the original posting information. > Does anyone have an updated way to do this? > > > > > -- Jerry (Gerald) Sedgewick Core Facility Director, Biomedical Image Processing Lab (BIPL) University of Minnesota, Department of Neuroscience 1-205 Hasselmo Hall 312 Church St. S.E. Minneapolis, MN 55455 612-624-6607 [hidden email] http://www.bipl.umn.edu Author: "Scientific Imaging with Photoshop: Methods, Measurement and Output." Rawlight.com (dba "Sedgewick Initiatives") 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 [hidden email] http://www.quickphotoshop.com http://www.rawlight.com http://www.jerrysedgewick.com | 
|   | 
				In reply to this post by Dan Osborn
			 
		I still swear by lens tissues over cotton swabs, mainly because the
 lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > | 
|   | 
		Hi All,
 First, before I used First Contact, I spoke with a person that is very prominent in the filter industry (trying to limit commercial conflict). That person said the polymer does not damage their coatings. I purchased the product. While reading this thread, I took out an emission filter and put a big old thumbprint on it. The polymer pulled it and all other impurities very nicely. You do have to be very careful near the edges of the mount. 0 commercial interest, but very happy. Best, Gary Laevsky, Ph.D. Imaging Application Specialist Andor Technology discover new ways of seeing [hidden email] Cell (774) 291 - 9992 Office (860) 290 - 9211 x219 Fax (860) 290 - 9566 Web: www.andor.com Please visit the following links for further information on the Andor microscopy systems http://www.andor.com/learning/movie_library/ please scroll down to the microscopy systems movie http://www.andor.com/microscopy_systems/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 7:00 PM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > | 
|   | 
				In reply to this post by Craig Brideau
			 
		You can get lint/fibre free cotton swabs, they are usually used for opthamology.
 Cheers Cam -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Thursday, 8 October 2009 10:00 AM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > No virus found in this incoming message. Checked by AVG - www.avg.com Version: 8.5.421 / Virus Database: 270.14.5/2418 - Release Date: 10/07/09 05:18:00 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. | 
|    | Boswell, Carl A - (cboswell) | 
|   | 
				In reply to this post by Gary Laevsky-2
			 
		Thanks, Gary.  My big concern is the worry over "soft" vs "hard" coatings, 
 and all the warnings associated with the former. Apparently older filters are more delicate than the newer ones. I'm told that what I have are soft coatings. I guess I'll take the cautious approach and try a small area first. Thanks to all for your input. If nothing else, it gives me courage to at least try something. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Gary Laevsky" <[hidden email]> To: <[hidden email]> Sent: Wednesday, October 07, 2009 4:34 PM Subject: Re: cleaning of filters Hi All, First, before I used First Contact, I spoke with a person that is very prominent in the filter industry (trying to limit commercial conflict). That person said the polymer does not damage their coatings. I purchased the product. While reading this thread, I took out an emission filter and put a big old thumbprint on it. The polymer pulled it and all other impurities very nicely. You do have to be very careful near the edges of the mount. 0 commercial interest, but very happy. Best, Gary Laevsky, Ph.D. Imaging Application Specialist Andor Technology discover new ways of seeing [hidden email] Cell (774) 291 - 9992 Office (860) 290 - 9211 x219 Fax (860) 290 - 9566 Web: www.andor.com Please visit the following links for further information on the Andor microscopy systems http://www.andor.com/learning/movie_library/ please scroll down to the microscopy systems movie http://www.andor.com/microscopy_systems/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 7:00 PM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item > numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not > have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and > a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and > doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > | 
|   | 
		Hi all,
 In my youth I was always advised never to clean dichroic and similar optical coated filters unless you really have to, and that dust and even minor fingerprints generally don't affect the fluorescence image quality - i.e. damaging the delicate filter coatings can degrade image quality more than the stuff you want to remove. Abrasive dust can be blown off, and I use a large hand puffer mostly now that filtered compressed air isn't available in the lab. Beware when using invertible 'aerosol can' dusters though as these can squirt the propellant all the filter surface [tried that with a Newvicon camera sensor years ago, didn't like it]. That said I have found sticky immersion oil and media type deposits all over such filter optics and this has required physical contact for removal. I tend to avoid water based products unless the deposit is likely to water soluble and perhaps fixed by solvents. I suppose for best results follow the makers instructions [Glen Spectra in this case]: From: http://www.glenspectra.co.uk/glen/filters/clean.htm "Cleaning of Optical Components Note: You follow this advice at your own risk. All optical elements are delicate and should be handled as carefully as possible. The glass and antireflective (AR) coated surfaces will be damaged by any contact, especially if abrasive particles have come into contact with the surface. In most cases, it is best to leave minor debris on the surface. Use of oil-free dry air or nitrogen under moderate pressure is the best tool for removing excessive debris from an optical surface. In the case that the contamination is not dislodged by the flow of gas, please use the following protocol for cleaning the part: 1. Clean the part using an absorbent towel such as Kimwipes, not lens paper. Use enough toweling so that solvents do not dissolve oils from your hands which can make their way through the toweling onto the coated surface. 2. Wet the towel with an anhydrous reagent grade ethanol. 3. The use of powder-free gloves will help to keep fingerprints off the part while cleaning. 4. Drag the trailing edge of the ethanol soaked Kimwipe across the surface of the component, moving in a single direction. A minimal amount of pressure can be applied while wiping. However, too much pressure will damage the component. 5. If the surface requires additional cleaning, always switch to a new Kimwipe before repeating the process. The purpose of the solvent is only to dissolve any adhesive contamination that is holding the debris on the surface. The towel needs to absorb both the excessive solvent and entrap the debris so that it can be removed from the surface. Surface coatings on interference filters and dichroics are typically less hard than the substrate. It is reasonable to expect that any cleaning will degrade the surface at an atomic level. Consideration should be given as to whether the contamination in question is more significant to the application than the damage that may result from cleaning the surface. In many cases, the AR coatings that are provided to give maximum light transmission amplify the appearance of contamination on the surface." So as usual, it's a clear case of yes.....and no. There are things like Newport cleaning tissues for 'drop & drag' cleaning coated optics [expensive but you don't need to use them much] http://search.newport.com/?q=cleaning that look a bit like glorified lens tissues. I tend to use clean soft tissues/cloths [depending on the problem & filter type/coatings, all with no 'rubbing'] with air blowing before and after. Regards Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages: http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Carl Boswell Sent: 08 October 2009 02:06 To: [hidden email] Subject: Re: cleaning of filters Thanks, Gary. My big concern is the worry over "soft" vs "hard" coatings, and all the warnings associated with the former. Apparently older filters are more delicate than the newer ones. I'm told that what I have are soft coatings. I guess I'll take the cautious approach and try a small area first. Thanks to all for your input. If nothing else, it gives me courage to at least try something. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Gary Laevsky" <[hidden email]> To: <[hidden email]> Sent: Wednesday, October 07, 2009 4:34 PM Subject: Re: cleaning of filters Hi All, First, before I used First Contact, I spoke with a person that is very prominent in the filter industry (trying to limit commercial conflict). That person said the polymer does not damage their coatings. I purchased the product. While reading this thread, I took out an emission filter and put a big old thumbprint on it. The polymer pulled it and all other impurities very nicely. You do have to be very careful near the edges of the mount. 0 commercial interest, but very happy. Best, Gary Laevsky, Ph.D. Imaging Application Specialist Andor Technology discover new ways of seeing [hidden email] Cell (774) 291 - 9992 Office (860) 290 - 9211 x219 Fax (860) 290 - 9566 Web: www.andor.com Please visit the following links for further information on the Andor microscopy systems http://www.andor.com/learning/movie_library/ please scroll down to the microscopy systems movie http://www.andor.com/microscopy_systems/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 7:00 PM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item > numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not > have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and > a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and > doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > | 
|   | 
				In reply to this post by Boswell, Carl A - (cboswell)
			 ** Commercial Response ** Please note that a protocol (including a video presentation) for cleaning all of Semrock filters is available at the following link:http://www.semrock.com/TechnicalInformation/TN_Cleaning/ Sincerely, Prashant Prashant Prabhat, Ph.D. Applications Specialist Semrock A Unit of IDEX Corporation 3625 Buffalo Road, Suite 6 Rochester NY 14624 Email: [hidden email] Phone: 585-594-7064 Toll Free: 866-SEMROCK Fax: 585-594-7095http://www.semrock.com 
 The Standard in Optical Filters for Biotech & Analytical Instrumentation Hundreds of Thousands of Ion Beam Sputtered filters delivered - extensive inventory now! The information contained in this message and any attachments may be privileged, confidential, and protected from disclosure. If the reader of this message is not the intended recipient, or any agent responsible for delivering this message to the intended recipient, you are hereby notified that any dissemination, distribution, or copying of this communication may be unlawful and therefore strictly prohibited. If you received this message in error, please reply to the message and delete it. 
 -----Original Message----- From: Confocal Microscopy List [[hidden email]] On Behalf Of Keith Morris Sent: Thursday, October 08, 2009 9:41 AM To: [hidden email] Subject: Re: cleaning of filters Hi all, In my youth I was always advised never to clean dichroic and similar optical coated filters unless you really have to, and that dust and even minor fingerprints generally don't affect the fluorescence image quality - i.e. damaging the delicate filter coatings can degrade image quality more than the stuff you want to remove. Abrasive dust can be blown off, and I use a large hand puffer mostly now that filtered compressed air isn't available in the lab. Beware when using invertible 'aerosol can' dusters though as these can squirt the propellant all the filter surface [tried that with a Newvicon camera sensor years ago, didn't like it]. That said I have found sticky immersion oil and media type deposits all over such filter optics and this has required physical contact for removal. I tend to avoid water based products unless the deposit is likely to water soluble and perhaps fixed by solvents. I suppose for best results follow the makers instructions [Glen Spectra in this case]: From: http://www.glenspectra.co.uk/glen/filters/clean.htm "Cleaning of Optical Components Note: You follow this advice at your own risk. All optical elements are delicate and should be handled as carefully as possible. The glass and antireflective (AR) coated surfaces will be damaged by any contact, especially if abrasive particles have come into contact with the surface. In most cases, it is best to leave minor debris on the surface. Use of oil-free dry air or nitrogen under moderate pressure is the best tool for removing excessive debris from an optical surface. In the case that the contamination is not dislodged by the flow of gas, please use the following protocol for cleaning the part: 1. Clean the part using an absorbent towel such as Kimwipes, not lens paper. Use enough toweling so that solvents do not dissolve oils from your hands which can make their way through the toweling onto the coated surface. 2. Wet the towel with an anhydrous reagent grade ethanol. 3. The use of powder-free gloves will help to keep fingerprints off the part while cleaning. 4. Drag the trailing edge of the ethanol soaked Kimwipe across the surface of the component, moving in a single direction. A minimal amount of pressure can be applied while wiping. However, too much pressure will damage the component. 5. If the surface requires additional cleaning, always switch to a new Kimwipe before repeating the process. The purpose of the solvent is only to dissolve any adhesive contamination that is holding the debris on the surface. The towel needs to absorb both the excessive solvent and entrap the debris so that it can be removed from the surface. Surface coatings on interference filters and dichroics are typically less hard than the substrate. It is reasonable to expect that any cleaning will degrade the surface at an atomic level. Consideration should be given as to whether the contamination in question is more significant to the application than the damage that may result from cleaning the surface. In many cases, the AR coatings that are provided to give maximum light transmission amplify the appearance of contamination on the surface." So as usual, it's a clear case of yes.....and no. There are things like Newport cleaning tissues for 'drop & drag' cleaning coated optics [expensive but you don't need to use them much]http://search.newport.com/?q=cleaning that look a bit like glorified lens tissues. I tend to use clean soft tissues/cloths [depending on the problem & filter type/coatings, all with no 'rubbing'] with air blowing before and after. Regards Keith --------------------------------------------------------------------------- Dr Keith J. Morris, Molecular Cytogenetics and Microscopy Core, Laboratory 00/069 and 00/070, The Wellcome Trust Centre for Human Genetics, Roosevelt Drive, Oxford OX3 7BN, United Kingdom. Telephone: +44 (0)1865 287568 Email: [hidden email] Web-pages:http://www.well.ox.ac.uk/molecular-cytogenetics-and-microscopy -----Original Message----- From: Confocal Microscopy List [[hidden email]] On Behalf Of Carl Boswell Sent: 08 October 2009 02:06 To: [hidden email] Subject: Re: cleaning of filters Thanks, Gary. My big concern is the worry over "soft" vs "hard" coatings, and all the warnings associated with the former. Apparently older filters are more delicate than the newer ones. I'm told that what I have are soft coatings. I guess I'll take the cautious approach and try a small area first. Thanks to all for your input. If nothing else, it gives me courage to at least try something. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Gary Laevsky" <[hidden email]> To: <[hidden email]> Sent: Wednesday, October 07, 2009 4:34 PM Subject: Re: cleaning of filters 
 Hi All, First, before I used First Contact, I spoke with a person that is very prominent in the filter industry (trying to limit commercial conflict). That person said the polymer does not damage their coatings. I purchased the product. While reading this thread, I took out an emission filter and put a big old thumbprint on it. The polymer pulled it and all other impurities very nicely. You do have to be very careful near the edges of the mount. 0 commercial interest, but very happy. Best, 
 
 Gary Laevsky, Ph.D. Imaging Application Specialist 
 
 Andor Technology discover new ways of seeing 
 
 Cell (774) 291 - 9992 Office (860) 290 - 9211 x219 Fax (860) 290 - 9566 Web:<A href="outbind://56/www.andor.com">www.andor.com 
 
 Please visit the following links for further information on the Andor microscopy systems 
 http://www.andor.com/learning/movie_library/ please scroll down to the microscopy systems movie 
 http://www.andor.com/microscopy_systems/default.aspx 
 -----Original Message----- From: Confocal Microscopy List [[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 7:00 PM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig 
 On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item > numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after cleaning, but > usually a circular swabbing motion from the inside out can reduce this. > There is also a fine polyester cleaning wipe used by some, but I do not > have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web:<A href="outbind://56/www.omegafilters.com">www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List [[hidden email]] > On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton together and > a > glue that attaches the cotton to the stick. Either of these could dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > >http://www.puritanmedproducts.com/search/search_4.asp?id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and > doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton fibers. > In my experience those can be blown away with a stream of pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, and it >> > is OK to use a good breath of air on the filter surface and either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > * * * * * * * * * * * * * This e-mail is intended solely for the intended recipient or recipients. If this e-mail is addressed to you in error or you otherwise receive this e-mail in error, please advise the sender, do not read, print, forward or save this e-mail, and promptly delete and destroy all copies of this e-mail. This email may contain information that is confidential, proprietary or secret and should be treated as confidential by all recipients. This e-mail may also be a confidential attorney-client communication, contain attorney work product, or otherwise be privileged and exempt from disclosure. If there is a confidentiality or non-disclosure agreement or protective order covering any information contained in this e-mail, such information shall be treated as confidential and subject to restriction on disclosure and use in accordance with such agreement or order, and this notice shall constitute identification, labeling or marking of such information as confidential, proprietary or secret in accordance with such agreement or order. The term 'this e-mail' includes any and all attachments. * * * * * * * * * * * * * 
 | 
|    | Peter Gabriel Pitrone | 
|   | 
				In reply to this post by Boswell, Carl A - (cboswell)
			 
		Hello Carl,
 If worse comes to worse, you can always replace it with a hard coated filter. Soft coated filters are at best 80-90% (more like 60-80%) transmission efficiencies, while hard coated filters have 95-99% efficiencies. Shorter exposure times are always welcomed. I have no connections to Semrock, Chroma or Omega. I'm just blown away at the comparisons between the two technologies. Pete On Oct 8, 2009, at 3:06 AM, Carl Boswell wrote: Thanks, Gary. My big concern is the worry over "soft" vs "hard" coatings, and all the warnings associated with the former. Apparently older filters are more delicate than the newer ones. I'm told that what I have are soft coatings. I guess I'll take the cautious approach and try a small area first. Thanks to all for your input. If nothing else, it gives me courage to at least try something. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Gary Laevsky" <[hidden email]> To: <[hidden email]> Sent: Wednesday, October 07, 2009 4:34 PM Subject: Re: cleaning of filters Hi All, First, before I used First Contact, I spoke with a person that is very prominent in the filter industry (trying to limit commercial conflict). That person said the polymer does not damage their coatings. I purchased the product. While reading this thread, I took out an emission filter and put a big old thumbprint on it. The polymer pulled it and all other impurities very nicely. You do have to be very careful near the edges of the mount. 0 commercial interest, but very happy. Best, Gary Laevsky, Ph.D. Imaging Application Specialist Andor Technology discover new ways of seeing [hidden email] Cell (774) 291 - 9992 Office (860) 290 - 9211 x219 Fax (860) 290 - 9566 Web: www.andor.com Please visit the following links for further information on the Andor microscopy systems http://www.andor.com/learning/movie_library/ please scroll down to the microscopy systems movie http://www.andor.com/microscopy_systems/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Wednesday, October 07, 2009 7:00 PM To: [hidden email] Subject: Re: cleaning of filters I still swear by lens tissues over cotton swabs, mainly because the lens tissues won't leave fibers behind. Most of the major optics companies sell their 'house brand' of these and they're all usually fairly good. I've used Edmund Optics and Thorlabs brands with good success. Edmund even sells large 'sheet-sized' pieces in packs. Craig On Wed, Oct 7, 2009 at 2:40 PM, Dan Osborn <[hidden email]> wrote: > Martin/ Deron, > We use several of the Puritan products on protected coatings and > mirrors: the same product as Deron listed, and 806 and 826 WC item > numbers. > The 869 was called out as the preferred product. And it true about the > fibers, there is tendency for some to be left behind after > cleaning, but > usually a circular swabbing motion from the inside out can reduce > this. > There is also a fine polyester cleaning wipe used by some, but I do > not have > the catalog number on that. > Best, > Dan > > > Dan Osborn > Product Marketing Manager > > Omega Optical, Inc. > Delta Campus > Omega Drive > Brattleboro, VT 05301 > Phone: Direct line: (802) 251-7305 or Toll Free: (866)-488-1064 > Fax: 802-254-3937 > Email: [hidden email] > Web: www.omegafilters.com > > > > > > Celebrating 40 Years > > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On > Behalf Of Deron Walters > Sent: Wednesday, October 07, 2009 4:02 PM > To: [hidden email] > Subject: Re: cleaning of filters > > The most common Q tips have both a binder to hold the cotton > together and a > glue that attaches the cotton to the stick. Either of these could > dissolve > in solvents and redeposit on optics. However, there is a glue-free, > binder-free swab that you can special order, the 869-WC from > Puritan: > > http://www.puritanmedproducts.com/search/search_4.asp? > id=336&item=869-WC > > I've used these with success on antireflection-coated singlets and > doublets, > although I haven't tested them on exposed filter coatings. > One objection to these is that (lacking binder) they shed cotton > fibers. > In my experience those can be blown away with a stream of > pressurized air. > > No financial interest. > > Deron Walters > R&D Scientist, Physics > Asylum Research > > >> On Behalf Of Martin Wessendorf >> >> Dan Osborn wrote: >> >> > However, some grime does come off better in aqueous solvents, >> and it >> > is OK to use a good breath of air on the filter surface and >> either a >> > Qtip or wipe to clean it. A water dampened Qtip or cloth followed > by >> > a dry wipe should be fine as well. >> >> I had always heard that Q-tips have starch in them as a binder, and > were >> thus unsuitable for optics--that we should make our own with cotton > wool >> and an applicator stick. Not so? > | 