cmos camera: appearance of boxes/bubbles over images

classic Classic list List threaded Threaded
♦
♦
13 messages Options
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Images can be seen here,
https://imgur.com/a/ff7aePj

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Do you see these features after some period of observing with the binocs (your eyes) and not using the camera? If so, I'd suspect your mounting medium.

Phil
-------------
Philip Oshel    
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Laurel Ballantyne <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Friday,  10January, 2020 at 10:55
To: "[hidden email]" <[hidden email]>
Subject: cmos camera: appearance of boxes/bubbles over images

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Images can be seen here,
    https://imgur.com/a/ff7aePj
   
    Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!
   

weber weber
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I had this once. There was moisture between the sensor and the protective glass.
Hamamatsu have changed the camera.
Best Regards
Michael Delannoy Michael Delannoy
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello,
I have never seen boxes before, but if the artifact appears under the beam to grow then the section is probably still wet.  Otherwise perhaps a salt contamination.  Good luck.

Sincerely,
Michael Delannoy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne
Sent: Friday, January 10, 2020 10:45 AM
To: [hidden email]
Subject: cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Images can be seen here,
https://imgur.com/a/ff7aePj

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!
Thews, Elmar Thews, Elmar
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [External] cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Moin Moin Laurel,

definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish.

Tschüß
Elmar
53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN
-
Dr. Elmar Thews | Principal Scientist, DAS
PerkinElmer | For the Better
[hidden email]
Phone:  +49 40 307090 353
Schnackenburgallee 114, D 22525 Hamburg
www.perkinelmer.com
Follow us on facebook and twitter
Please consider the environment before printing this e-mail.
This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc.  If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments.  Thank you.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne
Sent: Freitag, 10. Januar 2020 16:45
To: [hidden email]
Subject: [External] cmos camera: appearance of boxes/bubbles over images


Use caution when opening links or attachments.

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e=  and include the link in your posting.
*****

Images can be seen here,
https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e=

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!

PerkinElmer Cellular Technologies Germany GmbH
Sitz Hamburg
Amtsgericht Hamburg, HRB 101440
Geschäftsführer: Dr. Jürgen Müller
Tanneke den Blaauwen Tanneke den Blaauwen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Test whether you see it also if you only have an object glass. We had the same with an hamamatsu camera and got a replacement. Condense in the camera. For the rest the camara is very good : >)

Tanneke


On 10 Jan 2020, at 17:05, Oshel, Philip Eugene <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Do you see these features after some period of observing with the binocs (your eyes) and not using the camera? If so, I'd suspect your mounting medium.

Phil
-------------
Philip Oshel
Imaging Facility Director
Biology Department
1304 Biosciences
1455 Calumet Ct.
Central Michigan University
Mt. Pleasant, MI 48859
989 774-3576 office

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> on behalf of Laurel Ballantyne <[hidden email]>
Reply-To: Confocal Microscopy List <[hidden email]>
Date: Friday,  10January, 2020 at 10:55
To: "[hidden email]" <[hidden email]>
Subject: cmos camera: appearance of boxes/bubbles over images

   *****
   To join, leave or search the confocal microscopy listserv, go to:
   http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
   Post images on http://www.imgur.com and include the link in your posting.
   *****

   Images can be seen here,
   https://imgur.com/a/ff7aePj

   Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!



Guardare meravigliarsi et tornar a guardare

Tanneke den Blaauwen
Assoc prof Bacterial Cell Biology

Visiting address:
SILS, UvA
Science Park 904, C3.108
1098 XH Amsterdam, NL
[hidden email]<mailto:[hidden email]>
Phone 0031-205253852

Mail address:
T. den Blaauwen
Bacterial Cell Biology
SILS
PO BOX 94215
1090 GE Amsterdam
The Netherlands






Brian Kain Brian Kain
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Michael Delannoy
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

If you rotate the camera while viewing live, do the bubbles move with the image?  If not, then the bubbles are on the camera, and likely condensation from a failed seal within the camera.  Best to contact your local camera or microscope rep and review for warranty replacement.
Best,
Brian

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michael Delannoy
Sent: Friday, January 10, 2020 11:10 AM
To: [hidden email]
Subject: Re: cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=7E6gdo_iDOzdupnXYLvwKrPzxSvxUZiNErVr2L4McsQ&e= 
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=0qXBQZBk8J2vcXgO62cWyC3SdwGZ-GHyaumn7n5LcaU&e=  and include the link in your posting.
*****

Hello,
I have never seen boxes before, but if the artifact appears under the beam to grow then the section is probably still wet.  Otherwise perhaps a salt contamination.  Good luck.

Sincerely,
Michael Delannoy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne
Sent: Friday, January 10, 2020 10:45 AM
To: [hidden email]
Subject: cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=7E6gdo_iDOzdupnXYLvwKrPzxSvxUZiNErVr2L4McsQ&e= 
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=0qXBQZBk8J2vcXgO62cWyC3SdwGZ-GHyaumn7n5LcaU&e=  and include the link in your posting.
*****

Images can be seen here,
https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=fuLGaTLg3-tm9Cv-o53bE_3bZrx_TVqJ474o_cTXfWk&e= 

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

They're only there through the camera, not when looking through the eyepieces.
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks for the input, though quite sad.
Ho-chun LAI Ho-chun LAI
Reply | Threaded
Open this post in threaded view
♦
♦
|

回覆: [External] cmos camera: appearance of boxes/bubbles over images

In reply to this post by Thews, Elmar
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We did experienced once with a Zyla camera.
No the condensation won't go away and would need a factory refurbishing.
We then put a dehumidifier right next to it to prevent it from happening again.

Best regards,
John

________________________________
寄件者: Confocal Microscopy List <[hidden email]> 代表 Thews, Elmar <[hidden email]>
寄件日期: 2020年1月11日 0:12
收件者: [hidden email] <[hidden email]>
主旨: Re: [External] cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Moin Moin Laurel,

definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish.

Tschüß
Elmar
53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN
-
Dr. Elmar Thews | Principal Scientist, DAS
PerkinElmer | For the Better
[hidden email]
Phone:  +49 40 307090 353
Schnackenburgallee 114, D 22525 Hamburg
www.perkinelmer.com<http://www.perkinelmer.com>
Follow us on facebook and twitter
Please consider the environment before printing this e-mail.
This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc.  If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments.  Thank you.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne
Sent: Freitag, 10. Januar 2020 16:45
To: [hidden email]
Subject: [External] cmos camera: appearance of boxes/bubbles over images


Use caution when opening links or attachments.

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e=  and include the link in your posting.
*****

Images can be seen here,
https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e=

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!

PerkinElmer Cellular Technologies Germany GmbH
Sitz Hamburg
Amtsgericht Hamburg, HRB 101440
Geschäftsführer: Dr. Jürgen Müller
Cammer, Michael Cammer, Michael
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [External] cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****



Back in the day, we had to periodically hook up our Princeton Instruments camera to a vacuum pump.

More recently, a seal on a sCMOS broke.  One picture of the condensation that built up as the camera cooled:  https://www.flickr.com/photos/mcammer/34501637964/ 

The camera company fixed it.

Cheers-
Michael C.

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ho-chun LAI
Sent: Friday, January 10, 2020 11:26 AM
To: [hidden email]
Subject: 回覆: [External] cmos camera: appearance of boxes/bubbles over images

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=Kgwx6PHyaUR_SY1skT2EqrqMCdOyWkfFrZFuQ7zMXHg&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=zt0l-oU-aHIprjMLccnxlRuUrl6edggGyF4mDssWQAc&e=  and include the link in your posting.
*****

We did experienced once with a Zyla camera.
No the condensation won't go away and would need a factory refurbishing.
We then put a dehumidifier right next to it to prevent it from happening again.

Best regards,
John

________________________________
寄件者: Confocal Microscopy List <[hidden email]> 代表 Thews, Elmar <[hidden email]>
寄件日期: 2020年1月11日 0:12
收件者: [hidden email] <[hidden email]>
主旨: Re: [External] cmos camera: appearance of boxes/bubbles over images

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=Kgwx6PHyaUR_SY1skT2EqrqMCdOyWkfFrZFuQ7zMXHg&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=zt0l-oU-aHIprjMLccnxlRuUrl6edggGyF4mDssWQAc&e=  and include the link in your posting.
*****

Moin Moin Laurel,

definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish.

Tschüß
Elmar
53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN
-
Dr. Elmar Thews | Principal Scientist, DAS PerkinElmer | For the Better [hidden email]
Phone:  +49 40 307090 353
Schnackenburgallee 114, D 22525 Hamburg
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.perkinelmer.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=JtXLgVEmr6nRS5HiJQYt_9BA5vAMXgoDdSKpBA9uHfg&e= <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.perkinelmer.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=JtXLgVEmr6nRS5HiJQYt_9BA5vAMXgoDdSKpBA9uHfg&e= > Follow us on facebook and twitter Please consider the environment before printing this e-mail.
This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc.  If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments.  Thank you.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne
Sent: Freitag, 10. Januar 2020 16:45
To: [hidden email]
Subject: [External] cmos camera: appearance of boxes/bubbles over images


Use caution when opening links or attachments.

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e=  and include the link in your posting.
*****

Images can be seen here,
https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e=

Has anyone seen/experienced this issue before?  Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles".  These "boxes" are the same size no matter the magnification used.  Any help/advice is greatly appreciated!

PerkinElmer Cellular Technologies Germany GmbH Sitz Hamburg Amtsgericht Hamburg, HRB 101440
Geschäftsführer: Dr. Jürgen Müller
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Brian,

When I rotate the camera, the squares/bubbles don't move.  Thanks for the info!
Laurel Ballantyne Laurel Ballantyne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [External] cmos camera: appearance of boxes/bubbles over images

In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks so much, everyone, for all the info!  We'll look into sending the camera back for repair.