Laurel Ballantyne |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Images can be seen here, https://imgur.com/a/ff7aePj Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! |
|
Oshel, Philip Eugene |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Do you see these features after some period of observing with the binocs (your eyes) and not using the camera? If so, I'd suspect your mounting medium. Phil ------------- Philip Oshel Imaging Facility Director Biology Department 1304 Biosciences 1455 Calumet Ct. Central Michigan University Mt. Pleasant, MI 48859 989 774-3576 office -----Original Message----- From: Confocal Microscopy List <[hidden email]> on behalf of Laurel Ballantyne <[hidden email]> Reply-To: Confocal Microscopy List <[hidden email]> Date: Friday, 10January, 2020 at 10:55 To: "[hidden email]" <[hidden email]> Subject: cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Images can be seen here, https://imgur.com/a/ff7aePj Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I had this once. There was moisture between the sensor and the protective glass. Hamamatsu have changed the camera. Best Regards |
|
Michael Delannoy |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello, I have never seen boxes before, but if the artifact appears under the beam to grow then the section is probably still wet. Otherwise perhaps a salt contamination. Good luck. Sincerely, Michael Delannoy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne Sent: Friday, January 10, 2020 10:45 AM To: [hidden email] Subject: cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Images can be seen here, https://imgur.com/a/ff7aePj Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! |
|
Thews, Elmar |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Moin Moin Laurel, definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish. Tschüß Elmar 53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN - Dr. Elmar Thews | Principal Scientist, DAS PerkinElmer | For the Better [hidden email] Phone: +49 40 307090 353 Schnackenburgallee 114, D 22525 Hamburg www.perkinelmer.com Follow us on facebook and twitter Please consider the environment before printing this e-mail. This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc. If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments. Thank you. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne Sent: Freitag, 10. Januar 2020 16:45 To: [hidden email] Subject: [External] cmos camera: appearance of boxes/bubbles over images Use caution when opening links or attachments. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e= and include the link in your posting. ***** Images can be seen here, https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e= Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! PerkinElmer Cellular Technologies Germany GmbH Sitz Hamburg Amtsgericht Hamburg, HRB 101440 Geschäftsführer: Dr. Jürgen Müller |
|
Tanneke den Blaauwen |
|
|
In reply to this post by Oshel, Philip Eugene
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Test whether you see it also if you only have an object glass. We had the same with an hamamatsu camera and got a replacement. Condense in the camera. For the rest the camara is very good : >) Tanneke On 10 Jan 2020, at 17:05, Oshel, Philip Eugene <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Do you see these features after some period of observing with the binocs (your eyes) and not using the camera? If so, I'd suspect your mounting medium. Phil ------------- Philip Oshel Imaging Facility Director Biology Department 1304 Biosciences 1455 Calumet Ct. Central Michigan University Mt. Pleasant, MI 48859 989 774-3576 office -----Original Message----- From: Confocal Microscopy List <[hidden email]> on behalf of Laurel Ballantyne <[hidden email]> Reply-To: Confocal Microscopy List <[hidden email]> Date: Friday, 10January, 2020 at 10:55 To: "[hidden email]" <[hidden email]> Subject: cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Images can be seen here, https://imgur.com/a/ff7aePj Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! Guardare meravigliarsi et tornar a guardare Tanneke den Blaauwen Assoc prof Bacterial Cell Biology Visiting address: SILS, UvA Science Park 904, C3.108 1098 XH Amsterdam, NL [hidden email]<mailto:[hidden email]> Phone 0031-205253852 Mail address: T. den Blaauwen Bacterial Cell Biology SILS PO BOX 94215 1090 GE Amsterdam The Netherlands |
|
Brian Kain |
|
|
In reply to this post by Michael Delannoy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If you rotate the camera while viewing live, do the bubbles move with the image? If not, then the bubbles are on the camera, and likely condensation from a failed seal within the camera. Best to contact your local camera or microscope rep and review for warranty replacement. Best, Brian -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Michael Delannoy Sent: Friday, January 10, 2020 11:10 AM To: [hidden email] Subject: Re: cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=7E6gdo_iDOzdupnXYLvwKrPzxSvxUZiNErVr2L4McsQ&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=0qXBQZBk8J2vcXgO62cWyC3SdwGZ-GHyaumn7n5LcaU&e= and include the link in your posting. ***** Hello, I have never seen boxes before, but if the artifact appears under the beam to grow then the section is probably still wet. Otherwise perhaps a salt contamination. Good luck. Sincerely, Michael Delannoy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne Sent: Friday, January 10, 2020 10:45 AM To: [hidden email] Subject: cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=7E6gdo_iDOzdupnXYLvwKrPzxSvxUZiNErVr2L4McsQ&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=0qXBQZBk8J2vcXgO62cWyC3SdwGZ-GHyaumn7n5LcaU&e= and include the link in your posting. ***** Images can be seen here, https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIGaQ&c=7WMbxJudsInepJ54aTu5FB4kE34qljjw4wOb_YlLTpQ&r=gIFsZfJbmBb0FkHwBLdl_lx5F1ErV-NmoknCO-hDrUw&m=wUz-PScfcYnQlmJKViUrG8EMeNMYxj4kYDLPzDXq8E0&s=fuLGaTLg3-tm9Cv-o53bE_3bZrx_TVqJ474o_cTXfWk&e= Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! |
|
Laurel Ballantyne |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** They're only there through the camera, not when looking through the eyepieces. |
|
Laurel Ballantyne |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks for the input, though quite sad. |
|
Ho-chun LAI |
|
|
In reply to this post by Thews, Elmar
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We did experienced once with a Zyla camera. No the condensation won't go away and would need a factory refurbishing. We then put a dehumidifier right next to it to prevent it from happening again. Best regards, John ________________________________ 寄件者: Confocal Microscopy List <[hidden email]> 代表 Thews, Elmar <[hidden email]> 寄件日期: 2020年1月11日 0:12 收件者: [hidden email] <[hidden email]> 主旨: Re: [External] cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Moin Moin Laurel, definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish. Tschüß Elmar 53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN - Dr. Elmar Thews | Principal Scientist, DAS PerkinElmer | For the Better [hidden email] Phone: +49 40 307090 353 Schnackenburgallee 114, D 22525 Hamburg www.perkinelmer.com<http://www.perkinelmer.com> Follow us on facebook and twitter Please consider the environment before printing this e-mail. This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc. If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments. Thank you. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne Sent: Freitag, 10. Januar 2020 16:45 To: [hidden email] Subject: [External] cmos camera: appearance of boxes/bubbles over images Use caution when opening links or attachments. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e= and include the link in your posting. ***** Images can be seen here, https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e= Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! PerkinElmer Cellular Technologies Germany GmbH Sitz Hamburg Amtsgericht Hamburg, HRB 101440 Geschäftsführer: Dr. Jürgen Müller |
|
Cammer, Michael |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Back in the day, we had to periodically hook up our Princeton Instruments camera to a vacuum pump. More recently, a seal on a sCMOS broke. One picture of the condensation that built up as the camera cooled: https://www.flickr.com/photos/mcammer/34501637964/ The camera company fixed it. Cheers- Michael C. -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Ho-chun LAI Sent: Friday, January 10, 2020 11:26 AM To: [hidden email] Subject: 回覆: [External] cmos camera: appearance of boxes/bubbles over images [EXTERNAL] ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=Kgwx6PHyaUR_SY1skT2EqrqMCdOyWkfFrZFuQ7zMXHg&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=zt0l-oU-aHIprjMLccnxlRuUrl6edggGyF4mDssWQAc&e= and include the link in your posting. ***** We did experienced once with a Zyla camera. No the condensation won't go away and would need a factory refurbishing. We then put a dehumidifier right next to it to prevent it from happening again. Best regards, John ________________________________ 寄件者: Confocal Microscopy List <[hidden email]> 代表 Thews, Elmar <[hidden email]> 寄件日期: 2020年1月11日 0:12 收件者: [hidden email] <[hidden email]> 主旨: Re: [External] cmos camera: appearance of boxes/bubbles over images ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=Kgwx6PHyaUR_SY1skT2EqrqMCdOyWkfFrZFuQ7zMXHg&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=zt0l-oU-aHIprjMLccnxlRuUrl6edggGyF4mDssWQAc&e= and include the link in your posting. ***** Moin Moin Laurel, definitely broken seal in the camera with condensation on the inner surface of the protective window. If you leave the camera on for a few days it could vanish. Tschüß Elmar 53°35'15.55'' N; 9°54'21.35'' E, 25m above NHN - Dr. Elmar Thews | Principal Scientist, DAS PerkinElmer | For the Better [hidden email] Phone: +49 40 307090 353 Schnackenburgallee 114, D 22525 Hamburg https://urldefense.proofpoint.com/v2/url?u=http-3A__www.perkinelmer.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=JtXLgVEmr6nRS5HiJQYt_9BA5vAMXgoDdSKpBA9uHfg&e= <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.perkinelmer.com&d=DwIGaQ&c=j5oPpO0eBH1iio48DtsedeElZfc04rx3ExJHeIIZuCs&r=E0xNnPAQpUbDiPlC50tp7rW2nBkvV7fujQf0RknE5bU&m=FGWTfeXbsblWp0FlCGgPZaF4zNLCoJkEn7VKGZShjW0&s=JtXLgVEmr6nRS5HiJQYt_9BA5vAMXgoDdSKpBA9uHfg&e= > Follow us on facebook and twitter Please consider the environment before printing this e-mail. This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc. If you are not the intended recipient of this message, please inform the sender by replying to this e-mail or sending a message to the sender and destroy the message and any attachments. Thank you. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Laurel Ballantyne Sent: Freitag, 10. Januar 2020 16:45 To: [hidden email] Subject: [External] cmos camera: appearance of boxes/bubbles over images Use caution when opening links or attachments. ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=w2H93Sb9zb6KQr0QQubLWajuQkvGS6Tmc19vwiQwy3E&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=Ag0wlPciFMF1obUERPFPIWqXm6qnAwcipSIwIcRPK1E&e= and include the link in your posting. ***** Images can be seen here, https://urldefense.proofpoint.com/v2/url?u=https-3A__imgur.com_a_ff7aePj&d=DwIFaQ&c=3mGIBQKb1eS1YDHb7ax3kFkT2UAGBJ12p2lcECIjhXk&r=jHMoldduwAnzSMqmj26U9ExuDF2BnckbBFWEqk67Gxk&m=40F1muSN__EM9tyCncb6CZxGp4Y67giQ_hJyP3RtaZY&s=eVd-IN5uGYwcn4b-0gDZYiD0MIjKXrO7wWdNItRHCq8&e= Has anyone seen/experienced this issue before? Minutes into using the camera, these strange "boxes" appear in the center of the image and spread to cover the whole field, eventually turning into "bubbles". These "boxes" are the same size no matter the magnification used. Any help/advice is greatly appreciated! PerkinElmer Cellular Technologies Germany GmbH Sitz Hamburg Amtsgericht Hamburg, HRB 101440 Geschäftsführer: Dr. Jürgen Müller |
|
Laurel Ballantyne |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Brian, When I rotate the camera, the squares/bubbles don't move. Thanks for the info! |
|
Laurel Ballantyne |
|
|
In reply to this post by Laurel Ballantyne
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks so much, everyone, for all the info! We'll look into sending the camera back for repair. |