Hello all;
        Interesting volume localization and quantitation question for
you.
        One of my users wants to show that her marker moves to the
plasma membrane after stimulation. We have used Alexa 568 WGA on her
cells to label the membrane. Her channel protein is eGFP-labeled. Both
signals are not continuous, but are punctate. After stimulation, the
signal appears to move to the membrane, but the WGA and her eGFP do not
overlap, i.e., the signals are found in exclusive patches, as shown by
LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
WGA signal in MetaMorph (with a variety of values), made a binary mask,
and overlaid this on the GFP signal. What we see is the suggestion that
the channel protein is at the membrane in the same plane as the WGA, but
there is minimal overlap (as determined by line-scans). How to quantify
this? She likes my reference to the patches on a soccer ball...they are
obviously in the same spherical plane, but do not overlap. Would volume
rendering and modeling help?
        Thanks.
        Kathy



Kathryn Spencer, Ph.D.
The Scripps Research Institute
ICND 210
10550 N. Torrey Pines Road
La Jolla, CA  92037
(858) 784-8437
[hidden email]
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Kathy,

How about trying a Di dye?

They are more ubiquitous in the membrane and you could get them in
somewhat similar (DiI ex/em at 549/565)spectra.

No Commercial Interest

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Kathryn Spencer
Sent: Wednesday, August 06, 2008 4:44 PM
To: [hidden email]
Subject: co-planar localization

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello all;
        Interesting volume localization and quantitation question for
you.
        One of my users wants to show that her marker moves to the
plasma membrane after stimulation. We have used Alexa 568 WGA on her
cells to label the membrane. Her channel protein is eGFP-labeled. Both
signals are not continuous, but are punctate. After stimulation, the
signal appears to move to the membrane, but the WGA and her eGFP do not
overlap, i.e., the signals are found in exclusive patches, as shown by
LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
WGA signal in MetaMorph (with a variety of values), made a binary mask,
and overlaid this on the GFP signal. What we see is the suggestion that
the channel protein is at the membrane in the same plane as the WGA, but
there is minimal overlap (as determined by line-scans). How to quantify
this? She likes my reference to the patches on a soccer ball...they are
obviously in the same spherical plane, but do not overlap. Would volume
rendering and modeling help?
        Thanks.
        Kathy



Kathryn Spencer, Ph.D.
The Scripps Research Institute
ICND 210
10550 N. Torrey Pines Road
La Jolla, CA  92037
(858) 784-8437
[hidden email]
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Results were UGLY...WGA was nice and specific, but too punctate.
        What's new out your way? Are you liking Andor?
        Scripps is still plugging along. We have people moving in and
out of ICND, but all imaging types. We have someone coming in a few
weeks with her own multi-photon...finally I get to play!
        Kathy
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Gary Laevsky
Sent: Wednesday, August 06, 2008 4:23 PM
To: [hidden email]
Subject: Re: co-planar localization

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Kathy,

How about trying a Di dye?

They are more ubiquitous in the membrane and you could get them in
somewhat similar (DiI ex/em at 549/565)spectra.

No Commercial Interest

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell          (774) 291 - 9992
Office       (860) 290 - 9211 x219
Fax          (860) 290 - 9566
Web:       www.andor.com

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Kathryn Spencer
Sent: Wednesday, August 06, 2008 4:44 PM
To: [hidden email]
Subject: co-planar localization

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello all;
        Interesting volume localization and quantitation question for
you.
        One of my users wants to show that her marker moves to the
plasma membrane after stimulation. We have used Alexa 568 WGA on her
cells to label the membrane. Her channel protein is eGFP-labeled. Both
signals are not continuous, but are punctate. After stimulation, the
signal appears to move to the membrane, but the WGA and her eGFP do not
overlap, i.e., the signals are found in exclusive patches, as shown by
LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
WGA signal in MetaMorph (with a variety of values), made a binary mask,
and overlaid this on the GFP signal. What we see is the suggestion that
the channel protein is at the membrane in the same plane as the WGA, but
there is minimal overlap (as determined by line-scans). How to quantify
this? She likes my reference to the patches on a soccer ball...they are
obviously in the same spherical plane, but do not overlap. Would volume
rendering and modeling help?
        Thanks.
        Kathy



Kathryn Spencer, Ph.D.
The Scripps Research Institute
ICND 210
10550 N. Torrey Pines Road
La Jolla, CA  92037
(858) 784-8437
[hidden email]
Kevin Braeckmans Kevin Braeckmans
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

In reply to this post by kspencer007
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,

In my opinion there is not really a problem here, or at least no more of a
problem than there would be if the patches would be overlapping. The fact
that you see both types of patches in the same focal plane means obviously
that they are in the same plane within the microscope's resolution. If the
patches would be visually overlapping you would still have the same
uncertainty with regard to the z-position of both. It just looks a bit
nicer, that's all. Colocalisation based on the coincidence of colors in
still images is always limited by the spatial resolution.

Best regards,

Kevin


> -----Oorspronkelijk bericht-----
> Van: Confocal Microscopy List [mailto:[hidden email]]
> Namens Kathryn Spencer
> Verzonden: woensdag 6 augustus 2008 22:44
> Aan: [hidden email]
> Onderwerp: co-planar localization
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hello all;
> Interesting volume localization and quantitation question for
> you.
> One of my users wants to show that her marker moves to the
> plasma membrane after stimulation. We have used Alexa 568 WGA on her
> cells to label the membrane. Her channel protein is eGFP-labeled. Both
> signals are not continuous, but are punctate. After stimulation, the
> signal appears to move to the membrane, but the WGA and her eGFP do not
> overlap, i.e., the signals are found in exclusive patches, as shown by
> LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
> WGA signal in MetaMorph (with a variety of values), made a binary mask,
> and overlaid this on the GFP signal. What we see is the suggestion that
> the channel protein is at the membrane in the same plane as the WGA,
> but
> there is minimal overlap (as determined by line-scans). How to quantify
> this? She likes my reference to the patches on a soccer ball...they are
> obviously in the same spherical plane, but do not overlap. Would volume
> rendering and modeling help?
> Thanks.
> Kathy
>
>
>
> Kathryn Spencer, Ph.D.
> The Scripps Research Institute
> ICND 210
> 10550 N. Torrey Pines Road
> La Jolla, CA  92037
> (858) 784-8437
> [hidden email]
Grace Chojnowski Grace Chojnowski
Reply | Threaded
Open this post in threaded view
♦
♦
|

Darkfield Condenser

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have a Leica DMRE upright microscope with our Leica TCS SP2 confocal
system and are interested in sourcing a Dark Field condenser bottom for
it.
This part is no longer able to be supplied by Leica and I was
wondering/hoping that someone out there may have a spare one, or one
they longer need or want that they are willing to sell for a reasonable
price. The old Leica part number is 11 505 078.

Many thanks in advance.

Kind Regards,

Grace

 Grace Chojnowski
 Flow Cytometry and Confocal Microscopy Laboratory Manager
 Queensland Institute of Medical Research
 Brisbane, QLD, AUSTRALIA
 Email: [hidden email]
 Telephone 61-7-3362 0314.     Mob: 0414 956 752
 FAX 61-7-3362 0107
Bill Miller-3 Bill Miller-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Darkfield Condenser

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

you might want to look into  the Cytoviva system.
(http://www.cytoviva.com/main.htm) .. probably the ultimate darkfield
condenser..


Bill Miller


At 05:28 PM 8/7/2008 +1000, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>We have a Leica DMRE upright microscope with our Leica TCS SP2 confocal
>system and are interested in sourcing a Dark Field condenser bottom for
>it.
>This part is no longer able to be supplied by Leica and I was
>wondering/hoping that someone out there may have a spare one, or one
>they longer need or want that they are willing to sell for a reasonable
>price. The old Leica part number is 11 505 078.
>
>Many thanks in advance.
>
>Kind Regards,
>
>Grace
>
>  Grace Chojnowski
>  Flow Cytometry and Confocal Microscopy Laboratory Manager
>  Queensland Institute of Medical Research
>  Brisbane, QLD, AUSTRALIA
>  Email: [hidden email]
>  Telephone 61-7-3362 0314.     Mob: 0414 956 752
>  FAX 61-7-3362 0107
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

In reply to this post by Kevin Braeckmans
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Sounds to me like a TIRF microscope might help...

Cheers Mark Cannell

Kevin Braeckmans wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi,
>
> In my opinion there is not really a problem here, or at least no more of a
> problem than there would be if the patches would be overlapping. The fact
> that you see both types of patches in the same focal plane means obviously
> that they are in the same plane within the microscope's resolution. If the
> patches would be visually overlapping you would still have the same
> uncertainty with regard to the z-position of both. It just looks a bit
> nicer, that's all. Colocalisation based on the coincidence of colors in
> still images is always limited by the spatial resolution.
>
> Best regards,
>
> Kevin
>
>
>  
>> -----Oorspronkelijk bericht-----
>> Van: Confocal Microscopy List [mailto:[hidden email]]
>> Namens Kathryn Spencer
>> Verzonden: woensdag 6 augustus 2008 22:44
>> Aan: [hidden email]
>> Onderwerp: co-planar localization
>>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hello all;
>> Interesting volume localization and quantitation question for
>> you.
>> One of my users wants to show that her marker moves to the
>> plasma membrane after stimulation. We have used Alexa 568 WGA on her
>> cells to label the membrane. Her channel protein is eGFP-labeled. Both
>> signals are not continuous, but are punctate. After stimulation, the
>> signal appears to move to the membrane, but the WGA and her eGFP do not
>> overlap, i.e., the signals are found in exclusive patches, as shown by
>> LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
>> WGA signal in MetaMorph (with a variety of values), made a binary mask,
>> and overlaid this on the GFP signal. What we see is the suggestion that
>> the channel protein is at the membrane in the same plane as the WGA,
>> but
>> there is minimal overlap (as determined by line-scans). How to quantify
>> this? She likes my reference to the patches on a soccer ball...they are
>> obviously in the same spherical plane, but do not overlap. Would volume
>> rendering and modeling help?
>> Thanks.
>> Kathy
>>
>>
>>
>> Kathryn Spencer, Ph.D.
>> The Scripps Research Institute
>> ICND 210
>> 10550 N. Torrey Pines Road
>> La Jolla, CA  92037
>> (858) 784-8437
>> [hidden email]
>>    
Steve Hunter-2 Steve Hunter-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Darkfield Condenser

In reply to this post by Grace Chojnowski
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Grace,

Bill Miller is right it works extremely well.  Contact SciTech as they
are the local distributor for Cytoviva.

Cheers

Steve

-----Original Message-----
From: Grace Chojnowski [mailto:[hidden email]]
Sent: Thursday, 7 August 2008 5:28 PM
To: [hidden email]
Subject: Darkfield Condenser

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have a Leica DMRE upright microscope with our Leica TCS SP2 confocal
system and are interested in sourcing a Dark Field condenser bottom for
it.
This part is no longer able to be supplied by Leica and I was
wondering/hoping that someone out there may have a spare one, or one
they longer need or want that they are willing to sell for a reasonable
price. The old Leica part number is 11 505 078.

Many thanks in advance.

Kind Regards,

Grace

 Grace Chojnowski
 Flow Cytometry and Confocal Microscopy Laboratory Manager
 Queensland Institute of Medical Research
 Brisbane, QLD, AUSTRALIA
 Email: [hidden email]
 Telephone 61-7-3362 0314.     Mob: 0414 956 752
 FAX 61-7-3362 0107
ian gibbins ian gibbins
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

In reply to this post by kspencer007
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

What is that you want to quantify, Kathy?

If it is the spatial relations between the labeled structures in the
plane of the membrane, there is a range of spatial statistics that let
you decide if they are really associated or not (eg as compared with a
uniform or random distribution). Doing this in 3D can be a bit of a
trick, but there are several approaches that are not too difficult, as
long as you can get some quantitative data out  of your image sets
(typically x-y-z coordinates and grey-scale values of the features of
interest.)

In my experience, this proviso is surprisingly difficult to extract
easily from imaging programs that must have the data embedded within
them. I haven't looked at MetaMorph for a while, so I can't offer any
clues there. May be someone else can?

If this approach is what you are after, let me know, and I can give you
some references.

IAN


On Thursday, August 7, 2008, at 06:14  AM, Kathryn Spencer wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hello all;
> Interesting volume localization and quantitation question for
> you.
> One of my users wants to show that her marker moves to the
> plasma membrane after stimulation. We have used Alexa 568 WGA on her
> cells to label the membrane. Her channel protein is eGFP-labeled. Both
> signals are not continuous, but are punctate. After stimulation, the
> signal appears to move to the membrane, but the WGA and her eGFP do not
> overlap, i.e., the signals are found in exclusive patches, as shown by
> LSM confocal, 60x, zoom x2, 1.4NA in Z-stacks. I have thresholded the
> WGA signal in MetaMorph (with a variety of values), made a binary mask,
> and overlaid this on the GFP signal. What we see is the suggestion that
> the channel protein is at the membrane in the same plane as the WGA,
> but
> there is minimal overlap (as determined by line-scans). How to quantify
> this? She likes my reference to the patches on a soccer ball...they are
> obviously in the same spherical plane, but do not overlap. Would volume
> rendering and modeling help?
> Thanks.
> Kathy
>
>
>
> Kathryn Spencer, Ph.D.
> The Scripps Research Institute
> ICND 210
> 10550 N. Torrey Pines Road
> La Jolla, CA  92037
> (858) 784-8437
> [hidden email]
>
>

* * * * * * * * * * *
Prof Ian Gibbins
Anatomy & Histology
Flinders University
GPO Box 2100
Adelaide SA 5001
AUSTRALIA

[hidden email]
voice: +61-8-8204 5271
fax: +61-8-8277 0085

http://som.flinders.edu.au/FUSA/Anatomy/
http://www.flinders.edu.au/neuroscience
Beat Ludin Beat Ludin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: co-planar localization

In reply to this post by Kevin Braeckmans
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

At 08:39 07-08-2008, <[hidden email]> wrote:
>Colocalisation based on the coincidence of colors in
>still images is always limited by the spatial resolution.

Keep in mind though that resolution is commonly defined as the
minimum distance at which to objects of the same color in an image
can be discerned.
However, the relative position of objects of different colors can be
determined with much higher precision because their intensity
profiles can be analyzed individually (provided that the S/N is
sufficient and that chromatic abberrations of the system are accounted for).

Beat