|
Boswell, Carl A - (cboswell) |
|
|
Hi listers,
When one tries to assess colocalization, the question of where something is in "z" is confounded by the differences in focal point based on emission wavelength. I see no difficulty in assessing how much z shift there is, using subresolution beads labelled with different dyes. My problem comes with trying to apply any correction factor derived from the beads back to the original image data and thus to the colocalization determination. Any ideas? We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software that allows an adjustment in "z" position for a particular channel/wavelength. Thanks, Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 |
|
|
I recommend coated beads for this task. They have an inner core that
tests one wavelength, and an outer coating that tests the other. Find the difference in 'center of mass' between the solid core and the hollow sphere around it and you should have your wavelength displacement difference. If your optics are good this should be a non-issue anyway. It's always good to check for truth in advertising though, at least where 'achromatic' lenses are concerned... Craig On Thu, May 20, 2010 at 1:11 PM, Carl Boswell <[hidden email]> wrote: > Hi listers, > When one tries to assess colocalization, the question of where something is > in "z" is confounded by the differences in focal point based on emission > wavelength. I see no difficulty in assessing how much z shift there is, > using subresolution beads labelled with different dyes. My problem comes > with trying to apply any correction factor derived from the beads back to > the original image data and thus to the colocalization determination. Any > ideas? > > We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software > that allows an adjustment in "z" position for a particular > channel/wavelength. > > Thanks, > Carl > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 |
|
Martin Wessendorf-2 |
|
|
Dear Carl--
Always good to check the characteristics of the lens. However, I expect that if you're using a good quality lens that's clean, the lens not the problem. (If the lens is dirty--oil/mounting medium/etc on a dry lens or glycerol, etc, on an oil lens--it's a different story: that can introduce significant chromatic aberration.) Misalignment of the pinholes in the z-axis would produce z-shift. However, my recollection is that on the 510, you can only align one of the pinholes (--the meta pinhole) in z. Not sure why. When I've done 3-color mirror scans on a 510 with good quality oil-immersion lenses, they generally show some degree of misregistration in the z-axis--up to about a micron. I haven't noticed any place in the Handbook that deals with this issue--Jim, if you're out there, am I missing something? Good luck! Martin Wessendorf Craig Brideau wrote: > I recommend coated beads for this task. They have an inner core that > tests one wavelength, and an outer coating that tests the other. Find > the difference in 'center of mass' between the solid core and the > hollow sphere around it and you should have your wavelength > displacement difference. > If your optics are good this should be a non-issue anyway. It's > always good to check for truth in advertising though, at least where > 'achromatic' lenses are concerned... > > Craig > > > On Thu, May 20, 2010 at 1:11 PM, Carl Boswell > <[hidden email]> wrote: >> Hi listers, >> When one tries to assess colocalization, the question of where something is >> in "z" is confounded by the differences in focal point based on emission >> wavelength. I see no difficulty in assessing how much z shift there is, >> using subresolution beads labelled with different dyes. My problem comes >> with trying to apply any correction factor derived from the beads back to >> the original image data and thus to the colocalization determination. Any >> ideas? >> >> We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software >> that allows an adjustment in "z" position for a particular >> channel/wavelength. >> >> Thanks, >> Carl >> >> Carl A. Boswell, Ph.D. >> Molecular and Cellular Biology >> University of Arizona >> 520-954-7053 >> FAX 520-621-3709 -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
|
Glen MacDonald-2 |
|
|
Hi Carl,
This sounds like a classic issue with chromatic aberration. Which channel(s) exhibit the axial shift? Even in some planapo lenses, 405 excitation is shifted relative to the 488 line. Is this shift consistent across the field? If less in center then you can zoom in to avoid it. Glen On May 20, 2010, at 1:28 PM, Martin Wessendorf wrote: > Dear Carl-- > > Always good to check the characteristics of the lens. However, I expect that if you're using a good quality lens that's clean, the lens not the problem. (If the lens is dirty--oil/mounting medium/etc on a dry lens or glycerol, etc, on an oil lens--it's a different story: that can introduce significant chromatic aberration.) > > Misalignment of the pinholes in the z-axis would produce z-shift. However, my recollection is that on the 510, you can only align one of the pinholes (--the meta pinhole) in z. Not sure why. When I've done 3-color mirror scans on a 510 with good quality oil-immersion lenses, they generally show some degree of misregistration in the z-axis--up to about a micron. > > I haven't noticed any place in the Handbook that deals with this issue--Jim, if you're out there, am I missing something? > > Good luck! > > Martin Wessendorf > > Craig Brideau wrote: >> I recommend coated beads for this task. They have an inner core that >> tests one wavelength, and an outer coating that tests the other. Find >> the difference in 'center of mass' between the solid core and the >> hollow sphere around it and you should have your wavelength >> displacement difference. >> If your optics are good this should be a non-issue anyway. It's >> always good to check for truth in advertising though, at least where >> 'achromatic' lenses are concerned... >> Craig >> On Thu, May 20, 2010 at 1:11 PM, Carl Boswell >> <[hidden email]> wrote: >>> Hi listers, >>> When one tries to assess colocalization, the question of where something is >>> in "z" is confounded by the differences in focal point based on emission >>> wavelength. I see no difficulty in assessing how much z shift there is, >>> using subresolution beads labelled with different dyes. My problem comes >>> with trying to apply any correction factor derived from the beads back to >>> the original image data and thus to the colocalization determination. Any >>> ideas? >>> >>> We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software >>> that allows an adjustment in "z" position for a particular >>> channel/wavelength. >>> >>> Thanks, >>> Carl >>> >>> Carl A. Boswell, Ph.D. >>> Molecular and Cellular Biology >>> University of Arizona >>> 520-954-7053 >>> FAX 520-621-3709 > > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 e-mail: [hidden email] |
|
|
In reply to this post by Boswell, Carl A - (cboswell)
Any subpixel shift is problematic with noisy data -I would use a little
deconvolution on each channel to improve S/N then use spline interpolation to shift the data by sub pixel amounts (whole pix shifts don't need interpolation). (I'm assuming the raw data is properly sampled or else you can't justify a shift due to the aliasing problem). Don't know about other programs but it would be straight forward with proper data processing programs like IDL BUT it would be much better to use a well corrected objective in the first place of course... Hope this helps, Mark > Hi listers, > When one tries to assess colocalization, the question of where > something is in "z" is confounded by the differences in focal point > based on emission wavelength. I see no difficulty in assessing how > much z shift there is, using subresolution beads labelled with > different dyes. My problem comes with trying to apply any correction > factor derived from the beads back to the original image data and thus > to the colocalization determination. Any ideas? > > We'll be using a Zeiss 510 Meta, and I haven't seen anything in the > software that allows an adjustment in "z" position for a particular > channel/wavelength. > > Thanks, > Carl > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 |
|
|
In reply to this post by Boswell, Carl A - (cboswell)
Hi Carl,
I would have thought if you are using chromatically corrected objectives this shouldn't be a problem. Or am i missing something? Cheers Cam -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Carl Boswell Sent: Friday, 21 May 2010 5:12 AM To: [hidden email] Subject: colocalization and focus Hi listers, When one tries to assess colocalization, the question of where something is in "z" is confounded by the differences in focal point based on emission wavelength. I see no difficulty in assessing how much z shift there is, using subresolution beads labelled with different dyes. My problem comes with trying to apply any correction factor derived from the beads back to the original image data and thus to the colocalization determination. Any ideas? We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software that allows an adjustment in "z" position for a particular channel/wavelength. Thanks, Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. |
|
|
Other than missing the scroll wheel on my mouse and finding out after i
post that people have already suggested this :):):) -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cameron Nowell Sent: Friday, 21 May 2010 8:56 AM To: [hidden email] Subject: Re: colocalization and focus Hi Carl, I would have thought if you are using chromatically corrected objectives this shouldn't be a problem. Or am i missing something? Cheers Cam -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Carl Boswell Sent: Friday, 21 May 2010 5:12 AM To: [hidden email] Subject: colocalization and focus Hi listers, When one tries to assess colocalization, the question of where something is in "z" is confounded by the differences in focal point based on emission wavelength. I see no difficulty in assessing how much z shift there is, using subresolution beads labelled with different dyes. My problem comes with trying to apply any correction factor derived from the beads back to the original image data and thus to the colocalization determination. Any ideas? We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software that allows an adjustment in "z" position for a particular channel/wavelength. Thanks, Carl Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waive any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. |
|
|
In reply to this post by Glen MacDonald-2
Violet lasers are relatively recent. Microscope optics are still
playing catch-up to the shorter wavelengths so they don't perform as well for violet lines. If one of your sources is close to 400nm this can be a serious problem. On the other hand, if you are using fairly standard wavelengths that the objectives are supposed to be chromatically corrected for, then you shouldn't have any focal shift issues. The exception being, as mentioned by others, if you are using mismatched media. Craig On Thu, May 20, 2010 at 2:43 PM, Glen MacDonald <[hidden email]> wrote: > Hi Carl, > This sounds like a classic issue with chromatic aberration. Which channel(s) exhibit the axial shift? Even in some planapo lenses, 405 excitation is shifted relative to the 488 line. Is this shift consistent across the field? If less in center then you can zoom in to avoid it. > > Glen > On May 20, 2010, at 1:28 PM, Martin Wessendorf wrote: > >> Dear Carl-- >> >> Always good to check the characteristics of the lens. However, I expect that if you're using a good quality lens that's clean, the lens not the problem. (If the lens is dirty--oil/mounting medium/etc on a dry lens or glycerol, etc, on an oil lens--it's a different story: that can introduce significant chromatic aberration.) >> >> Misalignment of the pinholes in the z-axis would produce z-shift. However, my recollection is that on the 510, you can only align one of the pinholes (--the meta pinhole) in z. Not sure why. When I've done 3-color mirror scans on a 510 with good quality oil-immersion lenses, they generally show some degree of misregistration in the z-axis--up to about a micron. >> >> I haven't noticed any place in the Handbook that deals with this issue--Jim, if you're out there, am I missing something? >> >> Good luck! >> >> Martin Wessendorf >> >> Craig Brideau wrote: >>> I recommend coated beads for this task. They have an inner core that >>> tests one wavelength, and an outer coating that tests the other. Find >>> the difference in 'center of mass' between the solid core and the >>> hollow sphere around it and you should have your wavelength >>> displacement difference. >>> If your optics are good this should be a non-issue anyway. It's >>> always good to check for truth in advertising though, at least where >>> 'achromatic' lenses are concerned... >>> Craig >>> On Thu, May 20, 2010 at 1:11 PM, Carl Boswell >>> <[hidden email]> wrote: >>>> Hi listers, >>>> When one tries to assess colocalization, the question of where something is >>>> in "z" is confounded by the differences in focal point based on emission >>>> wavelength. I see no difficulty in assessing how much z shift there is, >>>> using subresolution beads labelled with different dyes. My problem comes >>>> with trying to apply any correction factor derived from the beads back to >>>> the original image data and thus to the colocalization determination. Any >>>> ideas? >>>> >>>> We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software >>>> that allows an adjustment in "z" position for a particular >>>> channel/wavelength. >>>> >>>> Thanks, >>>> Carl >>>> >>>> Carl A. Boswell, Ph.D. >>>> Molecular and Cellular Biology >>>> University of Arizona >>>> 520-954-7053 >>>> FAX 520-621-3709 >> >> -- >> Martin Wessendorf, Ph.D. office: (612) 626-0145 >> Assoc Prof, Dept Neuroscience lab: (612) 624-2991 >> University of Minnesota Preferred FAX: (612) 624-8118 >> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 >> Minneapolis, MN 55455 e-mail: [hidden email] > |
|
|
I dunno, when we checked for z-shift in green, red and far red
channels, even under optimal conditions with polychromatic beads we found chromatic abberation in the range of 200 to 400 nanometer, eg. for the red channel beeing above the other two (plan apo, NA 1.4). As expected, x-y abberations were much less, <100 nm. So, if you have a z-stack with distances of 250 nm (to fullfill Nyquist with the optimistic assumption that resolution in z is around 750 nm) it may well be worth to delete an image from opposite edges of the stacks. Carl, you didn't specify what you want to do, calculations or image overlays. For calculations it doesn't matter if it is sub pixel, you just define one color as standard and then apply the corrected positions of the other channels, e.g. before a distance measurement you tell the computer to move signals in a particular channel this and that far in x,y,z . For images, we didn't come up with something better than pixel-shift in z, sometimes in x,y. Good luck Steffen. At 01:42 21.05.2010, you wrote: >Violet lasers are relatively recent. Microscope optics are still >playing catch-up to the shorter wavelengths so they don't perform as >well for violet lines. If one of your sources is close to 400nm this >can be a serious problem. On the other hand, if you are using fairly >standard wavelengths that the objectives are supposed to be >chromatically corrected for, then you shouldn't have any focal shift >issues. The exception being, as mentioned by others, if you are using >mismatched media. > >Craig > >On Thu, May 20, 2010 at 2:43 PM, Glen MacDonald ><[hidden email]> wrote: > > Hi Carl, > > This sounds like a classic issue with chromatic > aberration. Which channel(s) exhibit the axial shift? Even in > some planapo lenses, 405 excitation is shifted relative to the 488 > line. Is this shift consistent across the field? If less in > center then you can zoom in to avoid it. > > > > Glen > > On May 20, 2010, at 1:28 PM, Martin Wessendorf wrote: > > > >> Dear Carl-- > >> > >> Always good to check the characteristics of the lens. However, > I expect that if you're using a good quality lens that's clean, the > lens not the problem. (If the lens is dirty--oil/mounting > medium/etc on a dry lens or glycerol, etc, on an oil lens--it's a > different story: that can introduce significant chromatic aberration.) > >> > >> Misalignment of the pinholes in the z-axis would produce > z-shift. However, my recollection is that on the 510, you can only > align one of the pinholes (--the meta pinhole) in z. Not sure > why. When I've done 3-color mirror scans on a 510 with good > quality oil-immersion lenses, they generally show some degree of > misregistration in the z-axis--up to about a micron. > >> > >> I haven't noticed any place in the Handbook that deals with this > issue--Jim, if you're out there, am I missing something? > >> > >> Good luck! > >> > >> Martin Wessendorf > >> > >> Craig Brideau wrote: > >>> I recommend coated beads for this task. They have an inner core that > >>> tests one wavelength, and an outer coating that tests the other. Find > >>> the difference in 'center of mass' between the solid core and the > >>> hollow sphere around it and you should have your wavelength > >>> displacement difference. > >>> If your optics are good this should be a non-issue anyway. It's > >>> always good to check for truth in advertising though, at least where > >>> 'achromatic' lenses are concerned... > >>> Craig > >>> On Thu, May 20, 2010 at 1:11 PM, Carl Boswell > >>> <[hidden email]> wrote: > >>>> Hi listers, > >>>> When one tries to assess colocalization, the question of where > something is > >>>> in "z" is confounded by the differences in focal point based on emission > >>>> wavelength. I see no difficulty in assessing how much z shift there is, > >>>> using subresolution beads labelled with different dyes. My > problem comes > >>>> with trying to apply any correction factor derived from the > beads back to > >>>> the original image data and thus to the colocalization > determination. Any > >>>> ideas? > >>>> > >>>> We'll be using a Zeiss 510 Meta, and I haven't seen anything > in the software > >>>> that allows an adjustment in "z" position for a particular > >>>> channel/wavelength. > >>>> > >>>> Thanks, > >>>> Carl > >>>> > >>>> Carl A. Boswell, Ph.D. > >>>> Molecular and Cellular Biology > >>>> University of Arizona > >>>> 520-954-7053 > >>>> FAX 520-621-3709 > >> > >> -- > >> Martin Wessendorf, Ph.D. office: (612) 626-0145 > >> Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > >> University of Minnesota Preferred FAX: (612) 624-8118 > >> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > >> Minneapolis, MN 55455 e-mail: [hidden email] > > |
|
|
In reply to this post by Boswell, Carl A - (cboswell)
I agree with previous responders that testing is definitely worth doing.
It is a good idea to use both the mirror slide and sub-resolution multicolor beads. The mirror slide will indicate the chromartic aberration and also the curvature of the field. When I was testing lenses on our confocal (Olympus), there was generally a reasonably small chromatic aberration (within their specifications, less than 250 nm) with the exception of the 405 nm laser, which was ~500 nm in some of the 'identical' 100x/1.4 oil immersion objectives, but more than that with some of the objectives. Also, the Plan correction may not be as good for the 405 nm wavelength. - A saw a fairly obvious curvature of the field in the 405 nm channel. The sub-resolution beads (I used Invitrogen Tetraspeck beads 100nm; no commercial interest) should indicate if the lateral registration of channels is uniform across the whole field of view, and reveal alignment problems. I was surprised in my tests to see that in a two-channel image the colocalization is perfect in one corner of the image, but in the opposite corner there was a lateral shift of several pixels. If aligning the microscope does not solve the problem completely, one could map these differences measured from the bead data stacks and apply affine transforms to warp one channel of the sample datasets. There was a post about this few years ago (quoted below); I never got around to implement it on our confocal. Stan Vitha Microscopy and Imaging Center Texas A&M University >Subject: Re: 3D registration of two channels >From: Kevin Braeckmans <[hidden email]> >Reply-To: Confocal Microscopy List <[hidden email]> >Date: Sun, 24 Sep 2006 07:10:22 +0200 >Content-Type: text/plain >You are right, it is also our experience that you can increase the overlay >accuracy by a more complex transformation than a simple translation. In our >case it is in a non-confocal application, but the principle stays the same. >By using a general affine transformation instead of a simple translation we >can increase the average overlay accuracy from approx. 0.5 micron to approx. >10 nm (this is in 2D). You could even use more complex transformations >(projective or even non-linear ones), but 10 nm is around the accuracy with >which we can determine the (Gaussian) centres of fluorescent beads. >(An affine transformation is a general combination of translation, scaling >along x, scaling along y, rotation, shear (i.e. transformation of angles), >and reflection along x and/or y.) >Before the actual experiments we take a dual channel image of Tetra Speck >beads. An image processing program we wrote in Matlab then calculates the >affine transformation parameters based on the (Gaussian) centres of the >beads. Finally, the dual channel movies are automatically overlayed using >those transformation parameters. >If you like you can send me a dual channel image of the Tetra Speck beads >(at least 4 beads in the image as widely spread as possible) and I could >check by how much you can improve the accuracy by using the more general >afiine transformation (maybe your confocal is already good as it is and a >translation is all you need). >Also, I think about 2 weeks ago someone on the list asked about the >registration accuracy of other confocal users. Maybe this person meanwhile >can share the information obtained from other confocalists what is the >overlay accuracy that is generally obtained on confocal microscopes? >Best regards, >Kevin >Kevin Braeckmans, Ph.D. >Lab. General Biochemistry & Physical Pharmacy >Ghent University >Harelbekestraat 72 >9000 Ghent >Belgium >Tel: +32 (0)9 264.80.78 >Fax: +32 (0)9 264.81.89 >E-mail: [hidden email] ============= On Thu, 20 May 2010 17:42:18 -0600, Craig Brideau <[hidden email]> wrote: >Violet lasers are relatively recent. Microscope optics are still >playing catch-up to the shorter wavelengths so they don't perform as >well for violet lines. If one of your sources is close to 400nm this >can be a serious problem. On the other hand, if you are using fairly >standard wavelengths that the objectives are supposed to be >chromatically corrected for, then you shouldn't have any focal shift >issues. The exception being, as mentioned by others, if you are using >mismatched media. > >Craig > >On Thu, May 20, 2010 at 2:43 PM, Glen MacDonald ><[hidden email]> wrote: >> Hi Carl, >> This sounds like a classic issue with chromatic aberration. Which excitation is shifted relative to the 488 line. Is this shift consistent across the field? If less in center then you can zoom in to avoid it. >> >> Glen >> On May 20, 2010, at 1:28 PM, Martin Wessendorf wrote: >> >>> Dear Carl-- >>> >>> Always good to check the characteristics of the lens. However, I expect that if you're using a good quality lens that's clean, the lens not the problem. (If the lens is dirty--oil/mounting medium/etc on a dry lens or glycerol, etc, on an oil lens--it's a different story: that can introduce significant chromatic aberration.) >>> >>> Misalignment of the pinholes in the z-axis would produce z-shift. However, my recollection is that on the 510, you can only align one of the pinholes (--the meta pinhole) in z. Not sure why. When I've done 3-color mirror scans on a 510 with good quality oil-immersion lenses, they generally show some degree of misregistration in the z-axis--up to about a micron. >>> >>> I haven't noticed any place in the Handbook that deals with this issue--Jim, if you're out there, am I missing something? >>> >>> Good luck! >>> >>> Martin Wessendorf >>> >>> Craig Brideau wrote: >>>> I recommend coated beads for this task. They have an inner core that >>>> tests one wavelength, and an outer coating that tests the other. Find >>>> the difference in 'center of mass' between the solid core and the >>>> hollow sphere around it and you should have your wavelength >>>> displacement difference. >>>> If your optics are good this should be a non-issue anyway. It's >>>> always good to check for truth in advertising though, at least where >>>> 'achromatic' lenses are concerned... >>>> Craig >>>> On Thu, May 20, 2010 at 1:11 PM, Carl Boswell >>>> <[hidden email]> wrote: >>>>> Hi listers, >>>>> When one tries to assess colocalization, the question of where >>>>> in "z" is confounded by the differences in focal point based on emission >>>>> wavelength. I see no difficulty in assessing how much z shift there is, >>>>> using subresolution beads labelled with different dyes. My problem comes >>>>> with trying to apply any correction factor derived from the beads back to >>>>> the original image data and thus to the colocalization determination. Any >>>>> ideas? >>>>> >>>>> We'll be using a Zeiss 510 Meta, and I haven't seen anything in the software >>>>> that allows an adjustment in "z" position for a particular >>>>> channel/wavelength. >>>>> >>>>> Thanks, >>>>> Carl >>>>> >>>>> Carl A. Boswell, Ph.D. >>>>> Molecular and Cellular Biology >>>>> University of Arizona >>>>> 520-954-7053 >>>>> FAX 520-621-3709 >>> >>> -- >>> Martin Wessendorf, Ph.D. office: (612) 626-0145 >>> Assoc Prof, Dept Neuroscience lab: (612) 624-2991 >>> University of Minnesota Preferred FAX: (612) 624-8118 >>> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 >>> Minneapolis, MN 55455 e-mail: [hidden email] >> |
| Free forum by Nabble | Edit this page |