confocal microscope efficiency measurement

classic Classic list List threaded Threaded
♦
♦
6 messages Options
fhenriques fhenriques
Reply | Threaded
Open this post in threaded view
♦
♦
|

confocal microscope efficiency measurement

 Dear all,

 First of all, hello everyone.

 I am new to the mailling list and the microscopy world.
 I'm a portughese student and am currently doing my master thesis at the
Instituto Gulbenkian de Ciência under the guidance of Nuno Moreno.

 We are trying to measure  the light capturing efficiency of a confocal
microscope but aren't sure of the methodology and were hoping you could shed
some light on it.

 The efficiency we want to measure is the fraction of photons emitted from a
light source that is detected in the final image, being: efficiency= np/ne,
where np is the number of detected photons and ne is the photons emitted by a
calibrated light source(LED).

 To calculate np we assumed there was no amplification noise, and used the mean
of the image M and it's variance V.
M=np*g and V=np*g^2.
Being g the average pixel intensity resulting from the detection of one photon,
which is proportional to the gain of the PMT and subsequent elecronics. And np
the average detected photons per pixel.

 Our doubt is in the calculation of ne.
Is each pixel a representation of all the light that reaches the lens during the
pixel dwell time?

 If so, ne should be the number of photons that reach the lens during the dwell
time.

 If not, if each pixel represents only a portion of the light that reaches the
lens during the dwell time, than ne should be a fraction of  the number of
photons that reach the lens during the dwell time.

 I have searched, but I am unable to find any kind of tutorial to help me
understand this concept. If somebody could point me in the direction of a good
source to understand how this should be done, I would be very grateful.

 Many thanks

 Francisco Henriques
Jason Swedlow Jason Swedlow
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: confocal microscope efficiency measurement

Hi Francisco-

Great project!

Actually, knowing the number of photons emitted from a fluorescent source (knowing in the sense of having a standard where the exact production of photons is "known") is very difficult.  Not impossible, but difficult.  You can get the efficiencies of the individual elements in the light path, and multiply them together.  This will give you an estimate of the efficiency.  John Murray did this in his chapter in Goldman and Spector's "Live Cell Imaging".

However, you can imagine coming up with a standard sample, and using that to compare different microscopes.  This was done recently:  search Pubmed for PMID: 18045334.  This might give you an idea for a standard sample you might use.

Good luck!

Cheers,

Jason

On Fri, Sep 11, 2009 at 4:42 PM, Francisco Henriques <[hidden email]> wrote:
 Dear all,

 First of all, hello everyone.

 I am new to the mailling list and the microscopy world.
 I'm a portughese student and am currently doing my master thesis at the
Instituto Gulbenkian de Ciência under the guidance of Nuno Moreno.

 We are trying to measure  the light capturing efficiency of a confocal
microscope but aren't sure of the methodology and were hoping you could shed
some light on it.

 The efficiency we want to measure is the fraction of photons emitted from a
light source that is detected in the final image, being: efficiency= np/ne,
where np is the number of detected photons and ne is the photons emitted by a
calibrated light source(LED).

 To calculate np we assumed there was no amplification noise, and used the mean
of the image M and it's variance V.
M=np*g and V=np*g^2.
Being g the average pixel intensity resulting from the detection of one photon,
which is proportional to the gain of the PMT and subsequent elecronics. And np
the average detected photons per pixel.

 Our doubt is in the calculation of ne.
Is each pixel a representation of all the light that reaches the lens during the
pixel dwell time?

 If so, ne should be the number of photons that reach the lens during the dwell
time.

 If not, if each pixel represents only a portion of the light that reaches the
lens during the dwell time, than ne should be a fraction of  the number of
photons that reach the lens during the dwell time.

 I have searched, but I am unable to find any kind of tutorial to help me
understand this concept. If somebody could point me in the direction of a good
source to understand how this should be done, I would be very grateful.

 Many thanks

 Francisco Henriques



--
**************************
Wellcome Trust Centre for Gene Regulation & Expression
College of Life Sciences
MSI/WTB/JBC Complex
University of Dundee
Dow Street
Dundee  DD1 5EH
United Kingdom

phone (01382) 385819
Intl phone:  44 1382 385819
FAX   (01382) 388072
email: [hidden email]

Lab Page: http://www.dundee.ac.uk/lifesciences/swedlow/
Open Microscopy Environment: http://openmicroscopy.org
**************************
Louis Villeneuve Louis Villeneuve
Reply | Threaded
Open this post in threaded view
♦
♦
|

Fixation with frozen Paraformaldehyde

In reply to this post by fhenriques

Bonjour à tous,

Can Paraformaldehyde 2-4% can be frozen for future use without lost of  fixation's efficiency?

Thanks for your help?

Louis




Francisco Henriques <[hidden email]>@LISTS.UMN.EDU
Envoyé par : Confocal Microscopy List <[hidden email]>

2009-09-11 11:42

Veuillez répondre à
Confocal Microscopy List <[hidden email]>

A
[hidden email]
cc
Objet
confocal microscope efficiency measurement







 Dear all,

First of all, hello everyone.

I am new to the mailling list and the microscopy world.
I'm a portughese student and am currently doing my master thesis at the
Instituto Gulbenkian de Ciência under the guidance of Nuno Moreno.

We are trying to measure  the light capturing efficiency of a confocal
microscope but aren't sure of the methodology and were hoping you could shed
some light on it.

The efficiency we want to measure is the fraction of photons emitted from a
light source that is detected in the final image, being: efficiency= np/ne,
where np is the number of detected photons and ne is the photons emitted by a
calibrated light source(LED).

To calculate np we assumed there was no amplification noise, and used the mean
of the image M and it's variance V.
M=np*g and V=np*g^2.
Being g the average pixel intensity resulting from the detection of one photon,
which is proportional to the gain of the PMT and subsequent elecronics. And np
the average detected photons per pixel.

Our doubt is in the calculation of ne.
Is each pixel a representation of all the light that reaches the lens during the
pixel dwell time?

If so, ne should be the number of photons that reach the lens during the dwell
time.

If not, if each pixel represents only a portion of the light that reaches the
lens during the dwell time, than ne should be a fraction of  the number of
photons that reach the lens during the dwell time.

I have searched, but I am unable to find any kind of tutorial to help me
understand this concept. If somebody could point me in the direction of a good
source to understand how this should be done, I would be very grateful.

Many thanks

Francisco Henriques



Ian Montgomery Ian Montgomery
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fixation with frozen Paraformaldehyde

Louis,

            Yes, do it all the time.

Ian.

 

Dr. Ian Montgomery,

Histotechnology,

I.B.L.S. Support Unit,

Thomson Building,

University of Glasgow,

Glasgow,

G12 8QQ.


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: 21 September 2009 14:35
To: [hidden email]
Subject: Fixation with frozen Paraformaldehyde

 


Bonjour à tous,

Can Paraformaldehyde 2-4% can be frozen for future use without lost of  fixation's efficiency?

Thanks for your help?

Louis



Francisco Henriques <[hidden email]>@LISTS.UMN.EDU
Envoyé par : Confocal Microscopy List <[hidden email]>

2009-09-11 11:42

Veuillez répondre à
Confocal Microscopy List <[hidden email]>

A

[hidden email]

cc

 

Objet

confocal microscope efficiency measurement

 

 

 






 Dear all,

First of all, hello everyone.

I am new to the mailling list and the microscopy world.
I'm a portughese student and am currently doing my master thesis at the
Instituto Gulbenkian de Ciência under the guidance of Nuno Moreno.

We are trying to measure  the light capturing efficiency of a confocal
microscope but aren't sure of the methodology and were hoping you could shed
some light on it.

The efficiency we want to measure is the fraction of photons emitted from a
light source that is detected in the final image, being: efficiency= np/ne,
where np is the number of detected photons and ne is the photons emitted by a
calibrated light source(LED).

To calculate np we assumed there was no amplification noise, and used the mean
of the image M and it's variance V.
M=np*g and V=np*g^2.
Being g the average pixel intensity resulting from the detection of one photon,
which is proportional to the gain of the PMT and subsequent elecronics. And np
the average detected photons per pixel.

Our doubt is in the calculation of ne.
Is each pixel a representation of all the light that reaches the lens during the
pixel dwell time?

If so, ne should be the number of photons that reach the lens during the dwell
time.

If not, if each pixel represents only a portion of the light that reaches the
lens during the dwell time, than ne should be a fraction of  the number of
photons that reach the lens during the dwell time.

I have searched, but I am unable to find any kind of tutorial to help me
understand this concept. If somebody could point me in the direction of a good
source to understand how this should be done, I would be very grateful.

Many thanks

Francisco Henriques


Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fixation with frozen Paraformaldehyde

In reply to this post by Louis Villeneuve
In my post-doc lab, we used to make the fixation solution fresh, use it, and then freeze it at -20C once.
Anything left over after thawing it once was discarded.
If you typically use a certain volume of fixation, you could freeze aliquots of the appropriate size for maximal use.
 
Cheers -
Lisa

---------------------------------------
Lisa Cameron
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Monday, September 21, 2009 9:35 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] Fixation with frozen Paraformaldehyde


Bonjour à tous,

Can Paraformaldehyde 2-4% can be frozen for future use without lost of  fixation's efficiency?

Thanks for your help?

Louis

The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
Sylvie Le Guyader-2 Sylvie Le Guyader-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Fixation with frozen Paraformaldehyde

I routinely freeze 10mL aliquots of PFA (freshly made from powder) 8% or 4%. I have used my stock without any problem for 2 years.

I thaw a working stock that I keep in the fridge for up to one month. I only warm up (if needed) the volume I need (so the aliquot stays in the fridge) and add it 2x to the cell culture with the same volume of warm medium. This perfectly preserves filopodia which are fragile structures.

 

Med vänlig hälsning / Best regards

 

Sylvie

 

@@@@@@@@@@@@@@@@@@@@@@@@

Sylvie Le Guyader

Dept of Biosciences and Nutrition

Karolinska Institutet

Novum

14157 Huddinge

Sweden

+46 (0)8 608 9240


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Cameron, Lisa
Sent: 21 September 2009 15:45
To: [hidden email]
Subject: Re: Fixation with frozen Paraformaldehyde

 

In my post-doc lab, we used to make the fixation solution fresh, use it, and then freeze it at -20C once.

Anything left over after thawing it once was discarded.

If you typically use a certain volume of fixation, you could freeze aliquots of the appropriate size for maximal use.

 

Cheers -

Lisa

---------------------------------------
Lisa Cameron
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Monday, September 21, 2009 9:35 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] Fixation with frozen Paraformaldehyde


Bonjour à tous,

Can Paraformaldehyde 2-4% can be frozen for future use without lost of  fixation's efficiency?

Thanks for your help?

Louis

 
 
The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.