Moulding, Dale |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? Cheers Dale |
|
Steffen Dietzel |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If you can use blue cellmask and Sir-700-Actin instead of phalloidin, you wouldn't have to. Else, instead of using A568 you could try with A555 (or Cy3) and A594 (or Cy3.5 or Texas Red) and maybe A647 (or Cy5) instead of A633. With 5 targets, I would go for a 5 color sample rather than relabeling. Steffen Am 16.02.2017 um 17:12 schrieb Dale Moulding: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? > Cheers > Dale > -- ------------------------------------------------------------ Steffen Dietzel, PD Dr. rer. nat Ludwig-Maximilians-Universität München Biomedical Center (BMC) Head of the Core Facility Bioimaging Großhaderner Straße 9 D-82152 Planegg-Martinsried Germany http://www.bioimaging.bmc.med.uni-muenchen.de |
|
Martin Wessendorf-2 |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Dale-- I know of no way to selectively quench the fluorescence of Alexa fluorochromes. However, back in the 1980s, we'd occasionally use elution-restaining to multi-stain tissue if we needed to localize two different antigens with primary antibodies raised in the same species. It involves treating the sample with potassium permanganate and sulphuric acid--not gentle, and background goes up. If you're stuck, it might provide an approach. A number of authors have used it to elute fluorescently labeled antibodies. Here's the reference: Tramu, G., A. Pillez, and J. Leonardelli (1978) An efficient method of antibody elution for the successive or simultaneous location of two antigens by immunocytochemistry. J. Histochem. Cytochem. 26: 322-324. Good luck! Martin On 2/16/2017 10:12 AM, Dale Moulding wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? > Cheers > Dale -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail:[hidden email] |
|
George McNamara |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Dale, These are chemically different classes, so do not count on one method working for all. While not specifically mentioned by you, Alexa Fluor 647 is a Cy5 analog, should be destroyed by the Gerdes et al (GE multiOmyx) method, see Highly multiplexed single-cell analysis of formalin-fixed, paraffin-embedded cancer tissue. Gerdes MJ, Sevinsky CJ, Sood A, Adak S, Bello MO, Bordwell A, Can A, Corwin A, Dinn S, Filkins RJ, Hollman D, Kamath V, Kaanumalle S, Kenny K, Larsen M, Lazare M, Li Q, Lowes C, McCulloch CC, McDonough E, Montalto MC, Pang Z, Rittscher J, Santamaria-Pang A, Sarachan BD, Seel ML, Seppo A, Shaikh K, Sui Y, Zhang J, Ginty F. Proc Natl Acad Sci U S A. 2013 Jul 16;110(29):11982-7. doi: 10.1073/pnas.1300136110. PMID: 23818604 and Gerdes patent. May be simpler to just strip the antibodies off (be sure to wear PPE - personal protective equipment -- while stripping antibodies). See slides 11 and 12 of https://digitalpathologyassociation.org/_data/files/2014_Pathology_Visions/PV14_Presentations/11_Day_2_Opening_Presentation_Levenson.pdf for a first round 10plex. Also mentioned at https://www.cacds.uh.edu/research/domain-applications/mapping-brain-tissue-alterations/ enjoy, George p.s. improving immunofluorescence on FFPE tissue sections - I encourage checking out Formaldehyde scavengers function as novel antigen retrieval agents. Vollert CT, Moree WJ, Gregory S, Bark SJ, Eriksen JL. Sci Rep. 2015 Nov 27;5:17322. doi: 10.1038/srep17322. PMID: 26612041 and their corresponding patent (US 9506928, http://www.freepatentsonline.com/9506928.html ), and to keep an eye on the company, www.teomics.com, for product(s). I do not have a financial connection to them (I met Craig V at JLabs@TMC events recently). They do not present quantitation on how much these 'formaldehyde scavengers' improve FFPE tissue sections antigen retrieval by -- my hope is 20x improement for 80% of epitopes ... which I think would greatly improve direct immunofluorescence, which in turn would enable multiplexing. Of course I had high hopes for an A.R. method a decade ago (PubMed 20025485, 15637333) which did not catch on. On 2/16/2017 10:12 AM, Dale Moulding wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? > Cheers > Dale -- George McNamara, PhD Houston, TX 77054 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75/ http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 |
|
Kelly Lundsten-2 |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Dale, We regularly do 5 color staining with Brilliant Violet 421, Brilliant Violet 510, AF488, AF594 (or AF555 if you choose) and AF647 or AF633 with a widefield scope and some minor filter changes. A lot of microscopists still don't know about the Brilliant Violet polymeric fluors since most people are using them only in flow cytometry. But, BV421 has an EC of 2.5 million and a QY of 65% in PBS with 450nm/450nm ex/em. You can see a 5 color spleen with our normal Olympus IX83. Though, we use direct conjugates for everything we can to overcome the species-dependent secondary detection issues you encounter with multiplexing microscopy. Outside of that, if you want to strip fluorescence, the only way to consistently kill any fluorophore is oxidation. A planar fluor must remain rigid to resonate energy. That's why some people will use peroxide incubations, to break a bond, release the rigidity of the hydrocarbon. Problem with any source of oxidation that can break a fluorophore, it'll also hurt cellular integrity/structure for the same reason, it's lack of selectivity. You have probably heard of people quenching endogenous fluorescence this same way. Just my two cents! Kelly Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding Sent: Thursday, February 16, 2017 8:13 AM To: [hidden email] Subject: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? Cheers Dale |
|
Romain Laine |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Dale, You might find that paper from Melike interesting: http://journals.plos.org/plosone/article?id=10.1371/journal.pone.0101772 They use sodium borohydride (NaBH4) to quench fluorescence before restaining with the same dye (AF647) but against different cellular structures. I hope this helps. Romain -- Dr. Romain Laine, PhD in Biophotonics Laser Analytics Group Department of Chemical Engineering and Biotechnology University of Cambridge New Museums Site Pembroke Street, Cambridge, CB2 3RA, UK T: (+44)1223330133 On Thu, Feb 16, 2017 at 5:54 PM, Kelly Lundsten <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Dale, > > We regularly do 5 color staining with Brilliant Violet 421, Brilliant > Violet 510, AF488, AF594 (or AF555 if you choose) and AF647 or AF633 with a > widefield scope and some minor filter changes. A lot of microscopists > still don't know about the Brilliant Violet polymeric fluors since most > people are using them only in flow cytometry. But, BV421 has an EC of 2.5 > million and a QY of 65% in PBS with 450nm/450nm ex/em. You can see a 5 > color spleen with our normal Olympus IX83. Though, we use direct > conjugates for everything we can to overcome the species-dependent > secondary detection issues you encounter with multiplexing microscopy. > > Outside of that, if you want to strip fluorescence, the only way to > consistently kill any fluorophore is oxidation. A planar fluor must remain > rigid to resonate energy. That's why some people will use peroxide > incubations, to break a bond, release the rigidity of the hydrocarbon. > Problem with any source of oxidation that can break a fluorophore, it'll > also hurt cellular integrity/structure for the same reason, it's lack of > selectivity. You have probably heard of people quenching endogenous > fluorescence this same way. > > Just my two cents! > Kelly > > > Kelly Lundsten > Business Segment Manager, Advanced Cytometry > Email: [hidden email] > Telephone: 773.633.4774 > Website: www.biolegend.com > BioLegend > 9727 Pacific Heights Blvd., San Diego, CA 92121, USA > > This message (including any attachments) may contain confidential, > proprietary, privileged and/or private information. The information is > intended to be for the use of the individual or entity designated above. If > you are not the intended recipient of this message, please notify the > sender immediately, and delete the message any attachments. Any disclosure, > distribution or other use of this message or any attachments by an > individual or entity other than the intended recipient is prohibited. > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Dale Moulding > Sent: Thursday, February 16, 2017 8:13 AM > To: [hidden email] > Subject: destroy alexa fluor fluorescence to allow restaining > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence > so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We > would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying > membranes and phalloidin binding? > Cheers > Dale > |
|
Rosemary.White |
|
|
In reply to this post by Kelly Lundsten-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kelly, Do you have any references detailing this method? This would be useful for us plant people as well, who have to deal with chloroplasts and other autofluorescence. thanks much, Rosemary Dr Rosemary White CSIRO Black Mountain GPO Box 1700 Canberra, ACT 2601 Australia Adjunct Prof, EH Graham Centre, CSU & Research School of Biology, ANU T 61 2 6246 5475 E [hidden email] On 17/2/17, 4:54 am, "Confocal Microscopy List on behalf of Kelly Lundsten" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Dale, We regularly do 5 color staining with Brilliant Violet 421, Brilliant Violet 510, AF488, AF594 (or AF555 if you choose) and AF647 or AF633 with a widefield scope and some minor filter changes. A lot of microscopists still don't know about the Brilliant Violet polymeric fluors since most people are using them only in flow cytometry. But, BV421 has an EC of 2.5 million and a QY of 65% in PBS with 450nm/450nm ex/em. You can see a 5 color spleen with our normal Olympus IX83. Though, we use direct conjugates for everything we can to overcome the species-dependent secondary detection issues you encounter with multiplexing microscopy. Outside of that, if you want to strip fluorescence, the only way to consistently kill any fluorophore is oxidation. A planar fluor must remain rigid to resonate energy. That's why some people will use peroxide incubations, to break a bond, release the rigidity of the hydrocarbon. Problem with any source of oxidation that can break a fluorophore, it'll also hurt cellular integrity/structure for the same reason, it's lack of selectivity. You have probably heard of people quenching endogenous fluorescence this same way. Just my two cents! Kelly Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding Sent: Thursday, February 16, 2017 8:13 AM To: [hidden email] Subject: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? Cheers Dale |
|
Kelly Lundsten-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rosemary, I totally concur on the sodium borohydride as well though I've never done that one myself. The peroxide is usually like 10-15% done on IHC-paraffin embedded tissue partly to reduce the autofluorescence but also to quench the endogenous peroxidase for people using HRP conjugates on the tissue later, so two birds one stone from the IHC perspective. I don't use the method myself so I was just googling for a protocol. I agree with Jason as well. All three of these methods H2O2, sodium borohydride and UV illumination or white light sources are all methods that Molecular Probes would recommend back in the day. UV and white light excitation are causing ROS production as the fluor excites and relaxes in the water environment, ROS are produced that go back to break the double bonds of the benzenes/hydrocarbons of the organic fluor structure. I don't know exactly how sodium borohydride does it though, since it's not an oxidizing agent. I suspect it quenches the fluors by making them no longer quantum efficient through a different reducing mechanism. Maybe Jason knows this one. Sorry I don't have much time to look into this more but here is a link to Abcam's blurb on it: http://www.abcam.com/kits/blocking-for-ihc . I remember the key always either being reduce the % H2O2 and lengthen the time or increase the %H2O2 to reduce the time.... all an optimization for the tissue and intensity of the signal you are hoping to destroy while doing minimal damage to the tissue. Good luck all! Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Thursday, February 16, 2017 1:35 PM To: [hidden email] Subject: Re: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kelly, Do you have any references detailing this method? This would be useful for us plant people as well, who have to deal with chloroplasts and other autofluorescence. thanks much, Rosemary Dr Rosemary White CSIRO Black Mountain GPO Box 1700 Canberra, ACT 2601 Australia Adjunct Prof, EH Graham Centre, CSU & Research School of Biology, ANU T 61 2 6246 5475 E [hidden email] On 17/2/17, 4:54 am, "Confocal Microscopy List on behalf of Kelly Lundsten" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Dale, We regularly do 5 color staining with Brilliant Violet 421, Brilliant Violet 510, AF488, AF594 (or AF555 if you choose) and AF647 or AF633 with a widefield scope and some minor filter changes. A lot of microscopists still don't know about the Brilliant Violet polymeric fluors since most people are using them only in flow cytometry. But, BV421 has an EC of 2.5 million and a QY of 65% in PBS with 450nm/450nm ex/em. You can see a 5 color spleen with our normal Olympus IX83. Though, we use direct conjugates for everything we can to overcome the species-dependent secondary detection issues you encounter with multiplexing microscopy. Outside of that, if you want to strip fluorescence, the only way to consistently kill any fluorophore is oxidation. A planar fluor must remain rigid to resonate energy. That's why some people will use peroxide incubations, to break a bond, release the rigidity of the hydrocarbon. Problem with any source of oxidation that can break a fluorophore, it'll also hurt cellular integrity/structure for the same reason, it's lack of selectivity. You have probably heard of people quenching endogenous fluorescence this same way. Just my two cents! Kelly Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding Sent: Thursday, February 16, 2017 8:13 AM To: [hidden email] Subject: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? Cheers Dale |
|
Jeffrey Carmichael |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL RESPONSE Hi Dale, I agree with Kelly that you can do this with 5 colors relatively easily *(although commercial responses should be labeled as such).* Below is a link to a slide deck on our website outlining the imaging scheme and filter sets to best separate these. BD Biosciences also offers the BV480 (like Aqua or CFP, but much, much brighter) which allows for distinct excitation and emission ranges for all 5 fluors. https://www.chroma.com/sites/ <goog_1687719810>default/files/5-CHANNEL% <goog_1687719810>20FLUORESCENCE%20IMAGING% <goog_1687719810> 20SIMPLIFIED%20-%20Reliable% <goog_1687719810>20Multiplexing%20for%20the% <goog_1687719810>20Non-Specialist_New.pdf Chroma does not profit from sales of any of the fluorophores, only from the filters. We also offer two different 5-band filter sets to accommodate this: https://www.chroma.com/products/sets/89903-et-bv421-bv480-af488-af568-af647-quinta-band-set https://www.chroma.com/products/sets/89904-et-405-445-514-561-640nm-laser-quinta-band-set I was initially skeptical that we could achieve good separation between BV480 and AF488, but with the right filter sets (specified in the slide deck) you can indeed. Good luck, Jeff *Jeff Carmichael* *Technical and Product Marketing Manager* *[hidden email] <[hidden email]>*Chroma Technology Corp. *an employee owned company* *10 Imtec Lane* *Bellows Falls, VT 05301* *802-428-2528 Office* *802-428-2528 Fax**800-824-7662 Toll Free* On Thu, Feb 16, 2017 at 11:12 AM, Dale Moulding <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence > so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We > would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying > membranes and phalloidin binding? > Cheers > Dale > |
|
|
In reply to this post by Kelly Lundsten-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** From a protocol we use for bleaching CNS before staining, maybe would work for stripping stains too: Incubate with NaIO4 (21.4 mg/ml in PBS) 15 min. followed by PBS wash Alternative bleaching solutions can be fresh NaBH4 or glycine followed by PBS wash =*===========================================================*= Michael Cammer, DART Microscopy Laboratory, NYU Langone Medical Center Cell: 914-309-3270 Office: Skirball 2nd Floor main office, back right http://ocs.med.nyu.edu/microscopy & http://microscopynotes.com/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kelly Lundsten Sent: Thursday, February 16, 2017 4:59 PM To: [hidden email] Subject: Re: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=ZKo8joI19MYnhE8lDWYvQv4y9TwX1v9wutSKYQ-hquk&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=jaSbIuPJZfYtZh5pwdlcbgJGVosDcO0_U6mn4nNKRFw&e= and include the link in your posting. ***** Hi Rosemary, I totally concur on the sodium borohydride as well though I've never done that one myself. The peroxide is usually like 10-15% done on IHC-paraffin embedded tissue partly to reduce the autofluorescence but also to quench the endogenous peroxidase for people using HRP conjugates on the tissue later, so two birds one stone from the IHC perspective. I don't use the method myself so I was just googling for a protocol. I agree with Jason as well. All three of these methods H2O2, sodium borohydride and UV illumination or white light sources are all methods that Molecular Probes would recommend back in the day. UV and white light excitation are causing ROS production as the fluor excites and relaxes in the water environment, ROS are produced that go back to break the double bonds of the benzenes/hydrocarbons of the organic fluor structure. I don't know exactly how sodium borohydride does it though, since it's not an oxidizing agent. I suspect it quenches the fluors by making them no longer quantum efficient through a different reducing mechanism. Maybe Jason knows this one. Sorry I don't have much time to look into this more but here is a link to Abcam's blurb on it: https://urldefense.proofpoint.com/v2/url?u=http-3A__www.abcam.com_kits_blocking-2Dfor-2Dihc&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=-oddrxaWstgi44kuXoZsDaU6GeZ1YB4nJW8600DXdEk&e= . I remember the key always either being reduce the % H2O2 and lengthen the time or increase the %H2O2 to reduce the time.... all an optimization for the tissue and intensity of the signal you are hoping to destroy while doing minimal damage to the tissue. Good luck all! Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Thursday, February 16, 2017 1:35 PM To: [hidden email] Subject: Re: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=ZKo8joI19MYnhE8lDWYvQv4y9TwX1v9wutSKYQ-hquk&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=jaSbIuPJZfYtZh5pwdlcbgJGVosDcO0_U6mn4nNKRFw&e= and include the link in your posting. ***** Hi Kelly, Do you have any references detailing this method? This would be useful for us plant people as well, who have to deal with chloroplasts and other autofluorescence. thanks much, Rosemary Dr Rosemary White CSIRO Black Mountain GPO Box 1700 Canberra, ACT 2601 Australia Adjunct Prof, EH Graham Centre, CSU & Research School of Biology, ANU T 61 2 6246 5475 E [hidden email] On 17/2/17, 4:54 am, "Confocal Microscopy List on behalf of Kelly Lundsten" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=ZKo8joI19MYnhE8lDWYvQv4y9TwX1v9wutSKYQ-hquk&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=jaSbIuPJZfYtZh5pwdlcbgJGVosDcO0_U6mn4nNKRFw&e= and include the link in your posting. ***** Hi Dale, We regularly do 5 color staining with Brilliant Violet 421, Brilliant Violet 510, AF488, AF594 (or AF555 if you choose) and AF647 or AF633 with a widefield scope and some minor filter changes. A lot of microscopists still don't know about the Brilliant Violet polymeric fluors since most people are using them only in flow cytometry. But, BV421 has an EC of 2.5 million and a QY of 65% in PBS with 450nm/450nm ex/em. You can see a 5 color spleen with our normal Olympus IX83. Though, we use direct conjugates for everything we can to overcome the species-dependent secondary detection issues you encounter with multiplexing microscopy. Outside of that, if you want to strip fluorescence, the only way to consistently kill any fluorophore is oxidation. A planar fluor must remain rigid to resonate energy. That's why some people will use peroxide incubations, to break a bond, release the rigidity of the hydrocarbon. Problem with any source of oxidation that can break a fluorophore, it'll also hurt cellular integrity/structure for the same reason, it's lack of selectivity. You have probably heard of people quenching endogenous fluorescence this same way. Just my two cents! Kelly Kelly Lundsten Business Segment Manager, Advanced Cytometry Email: [hidden email] Telephone: 773.633.4774 Website: www.biolegend.com BioLegend 9727 Pacific Heights Blvd., San Diego, CA 92121, USA This message (including any attachments) may contain confidential, proprietary, privileged and/or private information. The information is intended to be for the use of the individual or entity designated above. If you are not the intended recipient of this message, please notify the sender immediately, and delete the message any attachments. Any disclosure, distribution or other use of this message or any attachments by an individual or entity other than the intended recipient is prohibited. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding Sent: Thursday, February 16, 2017 8:13 AM To: [hidden email] Subject: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=ZKo8joI19MYnhE8lDWYvQv4y9TwX1v9wutSKYQ-hquk&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DQIFaQ&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=tTSr0amUjZOeOx4X_0vQqaRXof4boefIoOxcusaWDdI&s=jaSbIuPJZfYtZh5pwdlcbgJGVosDcO0_U6mn4nNKRFw&e= and include the link in your posting. ***** Dear all, does anyone know of a way to irreversibly destroy alexa fluor fluorescence so a sample can be stained with a second set of dyes? We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We would then like to re-image the same samples with cellmask and phalloidin. Is there any way to kill alexa fluor fluorescence without destroying membranes and phalloidin binding? Cheers Dale ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
Jeffrey Carmichael |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL RESPONSE Dale, In case the link didn't go through properly, I'll try again: https://www.chroma.com/sites/default/files/5-CHANNEL%20FLUORESCENCE%20IMAGING%20SIMPLIFIED%20-%20Reliable%20Multiplexing%20for%20the%20Non-Specialist_New.pdf Jeff *Jeff Carmichael* *Technical and Product Marketing Manager* *[hidden email] <[hidden email]>*Chroma Technology Corp. *an employee owned company* *10 Imtec Lane* *Bellows Falls, VT 05301* *802-428-2528 Office* *802-428-2528 Fax**800-824-7662 Toll Free* On Thu, Feb 16, 2017 at 11:12 AM, Dale Moulding <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor fluorescence > so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. We > would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying > membranes and phalloidin binding? > Cheers > Dale > |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** PW CHANGE MICROSPIM1_ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jeffrey Carmichael Sent: Thursday, February 16, 2017 5:57 PM To: [hidden email] Subject: Re: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL RESPONSE Dale, In case the link didn't go through properly, I'll try again: https://www.chroma.com/sites/default/files/5-CHANNEL%20FLUORESCENCE%20IMAGING%20SIMPLIFIED%20-%20Reliable%20Multiplexing%20for%20the%20Non-Specialist_New.pdf Jeff *Jeff Carmichael* *Technical and Product Marketing Manager* *[hidden email] <[hidden email]>*Chroma Technology Corp. *an employee owned company* *10 Imtec Lane* *Bellows Falls, VT 05301* *802-428-2528 Office* *802-428-2528 Fax**800-824-7662 Toll Free* On Thu, Feb 16, 2017 at 11:12 AM, Dale Moulding <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all, > does anyone know of a way to irreversibly destroy alexa fluor > fluorescence so a sample can be stained with a second set of dyes? > We do whole mount staining for 3 antigens with Alexa 488, 568 and 633. > We would then like to re-image the same samples with cellmask and phalloidin. > Is there any way to kill alexa fluor fluorescence without destroying > membranes and phalloidin binding? > Cheers > Dale > |
|
Moulding, Dale |
|
|
In reply to this post by Moulding, Dale
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, wow, thanks for the extremely helpful replies. I neglected to mention we have Dapi in the mix, so are looking for 6 colour imaging. From the replies this looks achievable. We’re not committed to any specific fluorophores, so it seems simply using Cy3 & Cy5 will allow us to re-use those channels for the Cell mask and Phalloidin. I’d not considered the BV fluorophores, these certainly look useful, as are the various chemicals (NaBH4, NaIO4, Glycine!?!) so we have plenty of options. Thanks! Dale |
|
Jim Mansfield |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If you are just looking for a 5-plex, 6-color, staining, there are lots of publications about how to do that without having to bleach, restain, reimage and then deal with all of the imaging artifacts introduced! https://jitc.biomedcentral.com/articles/10.1186/s40425-015-0091-z http://www.sciencedirect.com/science/article/pii/S1046202314002837 http://www.nature.com/articles/srep09534?WT.feed_name=subjects_imaging-the-immune-system http://link.springer.com/protocol/10.1007/978-1-4939-6730-8_5 Plus, there are commercial systems which enable this kind of staining using autostainers. Much of the above is possible on Leica Bond RX stainers, and Ventana/Roche has recently announced a 5-plex assay on their stainers. The publications above have all been optimized for thin FFPE tissue sections, so they might have to be modified for your samples. Best, Jim (no commercial interest) Jim Mansfield [hidden email] +1-204-619-3350 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dale Moulding Sent: February 20, 2017 7:56 AM To: [hidden email] Subject: Re: destroy alexa fluor fluorescence to allow restaining ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all, wow, thanks for the extremely helpful replies. I neglected to mention we have Dapi in the mix, so are looking for 6 colour imaging. From the replies this looks achievable. We’re not committed to any specific fluorophores, so it seems simply using Cy3 & Cy5 will allow us to re-use those channels for the Cell mask and Phalloidin. I’d not considered the BV fluorophores, these certainly look useful, as are the various chemicals (NaBH4, NaIO4, Glycine!?!) so we have plenty of options. Thanks! Dale |