Hi,

Does anyone know of an embedding medium which can be used in place of
paraffin, to reduce the autofluoresence produced?

Is there any literature available related to this topic?

Thanks.
Cameron Nowell Cameron Nowell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

Howdy,
 
You could try polyester wax. It is a bit trickier to cut than paraffin since it has to be kept cold otherwise it will be too soft to cut. you will also need to float the sections onto ice cold water instead of warm.
 
In my experience i have not seen autofluorecnce due to paraffin wax, usually it is due to the tissue type used.
 
Cheers
 
 
Cam
 
 
 
Cameron J. Nowell
Microscpy Manager
Central Resource for Advanced Microscopy
Ludwig Insttue for Cancer Research
PO Box 2008
Royal Melbourne Hospital
Victoria, 3050
AUSTRALIA
 
Office: +61 3 9341 3155
Mobile: +61422882700
Fax: +61 3 9341 3104
 
http://www.ludwig.edu.au/branch/research/platform/microscopy.htm
 

________________________________

From: Confocal Microscopy List on behalf of samrina aslam
Sent: Thu 16/04/2009 6:30 PM
To: [hidden email]
Subject: embedding medium



Hi,

Does anyone know of an embedding medium which can be used in place of
paraffin, to reduce the autofluoresence produced?

Is there any literature available related to this topic?

Thanks.
Samrina Aslam Samrina Aslam
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

Dear Cameron,
 
Thank you for your message. 
 
Okay is there a good method to reduce the autofluoresence? I am looking at amyloid deposits in various organs. However the autofluoresence is the same for all of them. I have tried the pressure cooker method for antigen retrieval but that has not reduced the background at all.
 
Thanks.
 
Samrina
 

> Date: Thu, 16 Apr 2009 19:20:03 +1000
> From: [hidden email]
> Subject: Re: embedding medium
> To: [hidden email]
>
> Howdy,
>
> You could try polyester wax. It is a bit trickier to cut than paraffin since it has to be kept cold otherwise it will be too soft to cut. you will also need to float the sections onto ice cold water instead of warm.
>
> In my experience i have not seen autofluorecnce due to paraffin wax, usually it is due to the tissue type used.
>
> Cheers
>
>
> Cam
>
>
>
> Cameron J. Nowell
> Microscpy Manager
> Central Resource for Advanced Microscopy
> Ludwig Insttue for Cancer Research
> PO Box 2008
> Royal Melbourne Hospital
> Victoria, 3050
> AUSTRALIA
>
> Office: +61 3 9341 3155
> Mobile: +61422882700
> Fax: +61 3 9341 3104
>
> http://www.ludwig.edu.au/branch/research/platform/microscopy.htm
>
>
> ________________________________
>
> From: Confocal Microscopy List on behalf of samrina aslam
> Sent: Thu 16/04/2009 6:30 PM
> To: [hidden email]
> Subject: embedding medium
>
>
>
> Hi,
>
> Does anyone know of an embedding medium which can be used in place of
> paraffin, to reduce the autofluoresence produced?
>
> Is there any literature available related to this topic?
>
> Thanks.


" Upgrade to Internet Explorer 8 Optimised for MSN. " Download Now
Martin Wessendorf-2 Martin Wessendorf-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

In reply to this post by Samrina Aslam
samrina aslam wrote:

> Does anyone know of an embedding medium which can be used in place of
> paraffin, to reduce the autofluoresence produced?
>
> Is there any literature available related to this topic?

I'm unaware of paraffin causing autofluorescence.  The usual causes in
histochemistry are:

1)  Glutaraldehyde fixation, which produces high levels of autofluorescence.
2)  Formaldehyde fixation for extended periods of time--also can cause
autofluorescence
3)  Lipofuscin--an autofluorescent pigment which accumulates in
lysosomes with age.  It's more of a problem with older animals than
younger ones.

--Autofluorescence from formaldehyde or low levels of glutaraldehyde can
be treated with NaBH4.  See: "Reduction of background autofluorescence
in brain sections following immersion in sodium borohydride."  Clancy B.
and Cauller LJ. J. Neuroscience Methods. 83(2):97-102, 1998 Sep 1.

--Lipfuscin autofluorescence can be decreased with Cu++ ion or Sudan
Black.   "Reduction of lipofuscin-like autofluorescence in fluorescently
labeled tissue."  Schnell SA. Staines WA. Wessendorf MW.
J. Histochemistry & Cytochemistry. 47(6):719-30, 1999 Jun.

Good luck!

Martin Wessendorf
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **
Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

In reply to this post by Samrina Aslam
Re: embedding medium

G
Best,

Gary Laevsky, Ph.D.
Imaging Application Specialist
Andor Technology
(774) 291 - 9992

----- Original Message -----
From: Confocal Microscopy List <[hidden email]>
To: [hidden email] <[hidden email]>
Sent: Thu Apr 16 14:15:38 2009
Subject: Re: embedding medium

samrina aslam wrote:

> Does anyone know of an embedding medium which can be used in place of
> paraffin, to reduce the autofluoresence produced?
>
> Is there any literature available related to this topic?

I'm unaware of paraffin causing autofluorescence.  The usual causes in
histochemistry are:

1)  Glutaraldehyde fixation, which produces high levels of autofluorescence.
2)  Formaldehyde fixation for extended periods of time--also can cause
autofluorescence
3)  Lipofuscin--an autofluorescent pigment which accumulates in
lysosomes with age.  It's more of a problem with older animals than
younger ones.

--Autofluorescence from formaldehyde or low levels of glutaraldehyde can
be treated with NaBH4.  See: "Reduction of background autofluorescence
in brain sections following immersion in sodium borohydride."  Clancy B.
and Cauller LJ. J. Neuroscience Methods. 83(2):97-102, 1998 Sep 1.

--Lipfuscin autofluorescence can be decreased with Cu++ ion or Sudan
Black.   "Reduction of lipofuscin-like autofluorescence in fluorescently
labeled tissue."  Schnell SA. Staines WA. Wessendorf MW.
J. Histochemistry & Cytochemistry. 47(6):719-30, 1999 Jun.

Good luck!

Martin Wessendorf
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

Gary Laevsky-2 Gary Laevsky-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

Re: embedding medium

Sorry.  Son to Blackberry

 

Best,

 

Gary Laevsky, Ph.D.

Imaging Application Specialist

 

Andor Technology

discover new ways of seeing

 

[hidden email]

Cell         (774) 291 - 9992

Office       (860) 290 - 9211 x219

Fax          (860) 290 - 9566

Web:       www.andor.com


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gary Laevsky
Sent: Thursday, April 16, 2009 9:33 AM
To: [hidden email]
Subject: Re: embedding medium

 

G
Best,

Gary Laevsky, Ph.D.
Imaging Application Specialist
Andor Technology
(774) 291 - 9992

----- Original Message -----
From: Confocal Microscopy List <[hidden email]>
To: [hidden email] <[hidden email]>
Sent: Thu Apr 16 14:15:38 2009
Subject: Re: embedding medium

samrina aslam wrote:

> Does anyone know of an embedding medium which can be used in place of
> paraffin, to reduce the autofluoresence produced?
>
> Is there any literature available related to this topic?

I'm unaware of paraffin causing autofluorescence.  The usual causes in
histochemistry are:

1)  Glutaraldehyde fixation, which produces high levels of autofluorescence.
2)  Formaldehyde fixation for extended periods of time--also can cause
autofluorescence
3)  Lipofuscin--an autofluorescent pigment which accumulates in
lysosomes with age.  It's more of a problem with older animals than
younger ones.

--Autofluorescence from formaldehyde or low levels of glutaraldehyde can
be treated with NaBH4.  See: "Reduction of background autofluorescence
in brain sections following immersion in sodium borohydride."  Clancy B.
and Cauller LJ. J. Neuroscience Methods. 83(2):97-102, 1998 Sep 1.

--Lipfuscin autofluorescence can be decreased with Cu++ ion or Sudan
Black.   "Reduction of lipofuscin-like autofluorescence in fluorescently
labeled tissue."  Schnell SA. Staines WA. Wessendorf MW.
J. Histochemistry & Cytochemistry. 47(6):719-30, 1999 Jun.

Good luck!

Martin Wessendorf
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: embedding medium

In reply to this post by Martin Wessendorf-2
Most mineral oils are fluorescent, so paraffin wax might well be.  But normally it's removed.  Maybe the problem is insufficient removeal of the wax before mounting?

                                   Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Martin Wessendorf
Sent: Thursday, 16 April 2009 11:16 PM
To: [hidden email]
Subject: Re: embedding medium

samrina aslam wrote:

> Does anyone know of an embedding medium which can be used in place of
> paraffin, to reduce the autofluoresence produced?
>
> Is there any literature available related to this topic?

I'm unaware of paraffin causing autofluorescence.  The usual causes in histochemistry are:

1)  Glutaraldehyde fixation, which produces high levels of autofluorescence.
2)  Formaldehyde fixation for extended periods of time--also can cause autofluorescence
3)  Lipofuscin--an autofluorescent pigment which accumulates in lysosomes with age.  It's more of a problem with older animals than younger ones.

--Autofluorescence from formaldehyde or low levels of glutaraldehyde can be treated with NaBH4.  See: "Reduction of background autofluorescence in brain sections following immersion in sodium borohydride."  Clancy B.
and Cauller LJ. J. Neuroscience Methods. 83(2):97-102, 1998 Sep 1.

--Lipfuscin autofluorescence can be decreased with Cu++ ion or Sudan
Black.   "Reduction of lipofuscin-like autofluorescence in fluorescently
labeled tissue."  Schnell SA. Staines WA. Wessendorf MW.
J. Histochemistry & Cytochemistry. 47(6):719-30, 1999 Jun.

Good luck!

Martin Wessendorf
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455
**MY E-MAIL ADDRESS HAS CHANGED.  PLEASE USE [hidden email] **

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.59/2063 - Release Date: 16/04/2009 4:38 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.59/2063 - Release Date: 16/04/2009 4:38 PM