Xinyu Zhao-2 |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, everybody, We have a user trying to do confocal microscopy on flatworms, ( parasitic blood flukes with a size of 1mm x 1cm, Schistosoma mansoni, ). The goal sounds easy: imaging the nucleus of the worms. And I thought the common procedures for immunotaining tissues would apply, i.e., PFA fixation followed by Triton 100x permeablization before the staining. However, in the handful literatures available, people seemed to always use something else, such as an alcohol-formalin-glacial acetic acid solution. Can anybody explain why ? And has anybody tried just PFA and Triton 100X on these creatures? Your input will be very much appreciated. Xinyu Zhao (Jasmine) Biomedical Imaging Core Lab Department of Pathology and Lab Medicine School of Medicine University of Pennsylvania Tel: 215-898-6730 |
|
Julian Smith III |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Our worms are much smaller (1mm x 200µm), but we're successfully using 4%PFA in 1x PBS with 10% sucrose and use triton X-100 for permeabilization. If your worms can be anesthetized, this should work fine. Julian Xinyu Zhao wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi, everybody, > > We have a user trying to do confocal microscopy on flatworms, ( parasitic blood flukes with a size of 1mm x 1cm, Schistosoma mansoni, ). The goal sounds easy: imaging the nucleus of the worms. And I thought the common procedures for immunotaining tissues would apply, i.e., PFA fixation followed by Triton 100x permeablization before the staining. > > However, in the handful literatures available, people seemed to always use something else, such as an alcohol-formalin-glacial acetic acid solution. > > Can anybody explain why ? And has anybody tried just PFA and Triton 100X on these creatures? > > Your input will be very much appreciated. > > > > Xinyu Zhao (Jasmine) > Biomedical Imaging Core Lab > Department of Pathology and Lab Medicine > School of Medicine > University of Pennsylvania > Tel: 215-898-6730 > > -- Julian P.S. Smith III Director, Winthrop Microscopy Facility Dept. of Biology Winthrop University 520 Cherry Rd. Rock Hill, SC 29733 803-323-2111 x6427 (vox) 803-323-3448 (fax) 803-524-2347 (cell) |
|
Renato A. Mortara |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Xinyu, we were successful in fixing different Schistosoma mansoni stages with 3.5% formaldehyde in PBS (see Bahia et al, Parasitology , 133: 321-329 2006). Good luck ! Renato Renato A. Mortara Parasitology Division UNIFESP - Escola Paulista de Medicina Rua Botucatu, 862, 6th floor São Paulo, SP 04023-062 Brazil Phone: 55 11 5579-8306 Fax: 55 11 5571-1095 email: [hidden email] home page: www.ecb.epm.br/~ramortara -----Mensagem original----- De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Julian Smith III Enviada em: quarta-feira, 13 de agosto de 2008 15:16 Para: [hidden email] Assunto: Re: fixation of flatworms for confocal microscopy Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Our worms are much smaller (1mm x 200µm), but we're successfully using 4%PFA in 1x PBS with 10% sucrose and use triton X-100 for permeabilization. If your worms can be anesthetized, this should work fine. Julian Xinyu Zhao wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi, everybody, > > We have a user trying to do confocal microscopy on flatworms, ( parasitic blood flukes with a size of 1mm x 1cm, Schistosoma mansoni, ). The goal sounds easy: imaging the nucleus of the worms. And I thought the common procedures for immunotaining tissues would apply, i.e., PFA fixation followed by Triton 100x permeablization before the staining. > > However, in the handful literatures available, people seemed to always use something else, such as an alcohol-formalin-glacial acetic acid solution. > > Can anybody explain why ? And has anybody tried just PFA and Triton 100X on these creatures? > > Your input will be very much appreciated. > > > > Xinyu Zhao (Jasmine) > Biomedical Imaging Core Lab > Department of Pathology and Lab Medicine School of Medicine University > of Pennsylvania > Tel: 215-898-6730 > > -- Julian P.S. Smith III Director, Winthrop Microscopy Facility Dept. of Biology Winthrop University 520 Cherry Rd. Rock Hill, SC 29733 803-323-2111 x6427 (vox) 803-323-3448 (fax) 803-524-2347 (cell) No virus found in this incoming message. Checked by AVG - http://www.avg.com Version: 8.0.138 / Virus Database: 270.6.2/1609 - Release Date: 13/8/2008 06:43 |
|
Stephen C. Kempf |
|
|
In reply to this post by Julian Smith III
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal If permeabilization of your worms turns out to be a problem, it may help to do a dehydration-rehydration series. We regularly run our larvae through the following prior to antibody labeling, distllled H2O 30% ethanol 50% ethanol 70% ethanol 50% ethanol 30% ethanol distllled H2O PBS + 0.1% Triton + 0.1% NaN3 PBS + 0.1% Triton + 0.1% NaN3 For our larvae (300 um), 10 min/solution works well. For larger pieces of tissue, you may want to hold them longer in each solution. Solutions may be room temperature or cold We find that this procedure gives better results than just treatment with Triton. Steve On Aug 13, 2008, at 1:16 PM, Julian Smith III wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Our worms are much smaller (1mm x 200µm), but we're successfully > using 4%PFA in 1x PBS with 10% sucrose and use triton X-100 for > permeabilization. > If your worms can be anesthetized, this should work fine. > Julian > > Xinyu Zhao wrote: >> Search the CONFOCAL archive at >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >> Hi, everybody, >> We have a user trying to do confocal microscopy on flatworms, >> ( parasitic blood flukes with a size of 1mm x 1cm, Schistosoma >> mansoni, ). The goal sounds easy: imaging the nucleus of the >> worms. And I thought the common procedures for immunotaining >> tissues would apply, i.e., PFA fixation followed by Triton 100x >> permeablization before the staining. >> >> However, in the handful literatures available, people seemed to >> always use something else, such as an alcohol-formalin-glacial >> acetic acid solution. >> >> Can anybody explain why ? And has anybody tried just PFA and >> Triton 100X on these creatures? >> Your input will be very much appreciated. >> >> >> >> Xinyu Zhao (Jasmine) >> Biomedical Imaging Core Lab >> Department of Pathology and Lab Medicine >> School of Medicine >> University of Pennsylvania >> Tel: 215-898-6730 >> >> > > > -- > Julian P.S. Smith III > Director, Winthrop Microscopy Facility > Dept. of Biology > Winthrop University > 520 Cherry Rd. > Rock Hill, SC 29733 > > 803-323-2111 x6427 (vox) > 803-323-3448 (fax) > 803-524-2347 (cell) |
|
Pierre Nourdine Bouchelouche |
|
|
In reply to this post by Xinyu Zhao-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi everybody, we have just purchased an SP5 confocal trying to do calcium measurements in smooth muscle cells. Do anybody have any experience with calcium measuremenst with SP5. Thank you. Best regards Pierre Smooth Muscle Research Center Copenhagen Denmark ________________________________________ Fra: Confocal Microscopy List [[hidden email]] På vegne af Xinyu Zhao [[hidden email]] Sendt: 13. august 2008 20:00 Til: [hidden email] Emne: fixation of flatworms for confocal microscopy Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi, everybody, We have a user trying to do confocal microscopy on flatworms, ( parasitic blood flukes with a size of 1mm x 1cm, Schistosoma mansoni, ). The goal sounds easy: imaging the nucleus of the worms. And I thought the common procedures for immunotaining tissues would apply, i.e., PFA fixation followed by Triton 100x permeablization before the staining. However, in the handful literatures available, people seemed to always use something else, such as an alcohol-formalin-glacial acetic acid solution. Can anybody explain why ? And has anybody tried just PFA and Triton 100X on these creatures? Your input will be very much appreciated. Xinyu Zhao (Jasmine) Biomedical Imaging Core Lab Department of Pathology and Lab Medicine School of Medicine University of Pennsylvania Tel: 215-898-6730 |