graphic cards

classic Classic list List threaded Threaded
♦
♦
27 messages Options
12
Marc Thibault Marc Thibault
Reply | Threaded
Open this post in threaded view
♦
♦
|

graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I would like to know what are the best graphic cards to add in your PC for microscopic 3D image analysis and manipulation.

Mine is currently not up to the task.

Thanks

Marc

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather than your GPU.  The CPU does all the number crunching then hands the resulting data off to the GPU (video card) to display it.  There are a handful of programs that will actually use the GPU instead, but they are few and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your computer?

Craig


On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]> wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi all,

I would like to know what are the best graphic cards to add in your PC for microscopic 3D image analysis and manipulation.

Mine is currently not up to the task.

Thanks

Marc


Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
           I owuld like to disagree with Craig. There are now quite a few programs that use the graphics card to process 3D data in real time now. MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to name a few.
 
There is an enourmous ammount of choice out there in the graphics card market, but my recomendation (and the recomendation of many other people) is to buy the best GAMING card you can afford. While each graphics card manuafacturer has "Workstation" class cards that cost a heap of money, these cards oare not really any better than a good gaming card. So avoid things like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.
 
So you are going to need something like an Nvidia Geforce 280GTX or an ATI Radeon HD 4870. You will need at least 512MB of RAM on the card as well. Also these cards cost in the order of $500-600 US dollars.
 
Now slapping a highend card in a lower end computer is not going to fix things either. To match the card you will need to have a good CPU (dual coe is fine, there are very, very few programs coded for more than 2 cores currently) and the more system RAM the better. If you are playing with large data sets (eg 2000 images per 3D stack) you will need to go beyond the maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16 or 32 GB RAM into it.
 
Fast hardrives will always help a little bit but not too much.
 
Also don't worry about going for a two graphics cards running togther (Nvidia SLI or ATI CrossFire), as far as i know there are no modeling programs out there yet that can leaverage both cards.
 
Hope that helps you out, if you need any more help just let me know.
 
 
 
Cheers
 
 
Cam
 
 
 
Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
Actually most image display and manipulation software uses your CPU rather than your GPU.  The CPU does all the number crunching then hands the resulting data off to the GPU (video card) to display it.  There are a handful of programs that will actually use the GPU instead, but they are few and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]> wrote:


        Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

        Hi all,

        I would like to know what are the best graphic cards to add in your PC for microscopic 3D image analysis and manipulation.

        Mine is currently not up to the task.

        Thanks

        Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
 
I've learned that imaging software may be optimized for particular hardware, so the "best" specifications may not necessarily make your program run the fastest.  I'd call the manufacturer of the application and ask them what their recommended CPU and graphics card configuration is.
 
-Esteban


 
On Wed, Aug 13, 2008 at 3:44 PM, Nowell, Cameron <[hidden email]> wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
          I owuld like to disagree with Craig. There are now quite a few programs that use the graphics card to process 3D data in real time now. MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to name a few.

There is an enourmous ammount of choice out there in the graphics card market, but my recomendation (and the recomendation of many other people) is to buy the best GAMING card you can afford. While each graphics card manuafacturer has "Workstation" class cards that cost a heap of money, these cards oare not really any better than a good gaming card. So avoid things like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.

So you are going to need something like an Nvidia Geforce 280GTX or an ATI Radeon HD 4870. You will need at least 512MB of RAM on the card as well. Also these cards cost in the order of $500-600 US dollars.

Now slapping a highend card in a lower end computer is not going to fix things either. To match the card you will need to have a good CPU (dual coe is fine, there are very, very few programs coded for more than 2 cores currently) and the more system RAM the better. If you are playing with large data sets (eg 2000 images per 3D stack) you will need to go beyond the maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16 or 32 GB RAM into it.

Fast hardrives will always help a little bit but not too much.

Also don't worry about going for a two graphics cards running togther (Nvidia SLI or ATI CrossFire), as far as i know there are no modeling programs out there yet that can leaverage both cards.

Hope that helps you out, if you need any more help just let me know.



Cheers


Cam



Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather than your GPU.  The CPU does all the number crunching then hands the resulting data off to the GPU (video card) to display it.  There are a handful of programs that will actually use the GPU instead, but they are few and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]> wrote:


       Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

       Hi all,

       I would like to know what are the best graphic cards to add in your PC for microscopic 3D image analysis and manipulation.

       Mine is currently not up to the task.

       Thanks

       Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.



--
G. Esteban Fernandez, Ph.D.
Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO 65211

http://www.biotech.missouri.edu/mcc/

573-882-4895
573-884-9395 fax

Adrian Smith-6 Adrian Smith-6
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

On 14/08/2008, at 4:44 AM, Nowell, Cameron wrote:

>           I owuld like to disagree with Craig. There are now quite a  
> few programs that use the graphics card to process 3D data in real  
> time now. MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000,  
> Leica LAS AF just to name a few.
>
> There is an enourmous ammount of choice out there in the graphics  
> card market, but my recomendation (and the recomendation of many  
> other people) is to buy the best GAMING card you can afford. While  
> each graphics card manuafacturer has "Workstation" class cards that  
> cost a heap of money, these cards oare not really any better than a  
> good gaming card. So avoid things like the Nvidia Quadro, ATI  
> FirePro or Wildcat Studio type of cards.


Here is a recent "value" analysis of gaming cards...

http://www.dansdata.com/askdan00034.htm

Adrian
Michael C. Wussow Michael C. Wussow
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hello,

        I have to completely agree with Cam.  His comments are spot on and
what we at Bitplane recommend to our customers.  However, it is recommend
that you check with the company that you are getting the software from just
to make sure there are no incompatibilities between the software and the
graphics card.

If you want specific recommendations don't hesitate to contact us offline.

Mike

Bitplane Inc.
Michael C. Wussow
Vice President and General Manager Bitplane Inc.
 
Cell Phone:    651-336-4600
Fax:                 866-691-9112
Toll Free:       1-888-3D-BITPX (332-4879)
Visit Our Web Site At:  www.bitplane.com



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, August 13, 2008 3:45 PM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
           I owuld like to disagree with Craig. There are now quite a few
programs that use the graphics card to process 3D data in real time now.
MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to
name a few.
 
There is an enourmous ammount of choice out there in the graphics card
market, but my recomendation (and the recomendation of many other people) is
to buy the best GAMING card you can afford. While each graphics card
manuafacturer has "Workstation" class cards that cost a heap of money, these
cards oare not really any better than a good gaming card. So avoid things
like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.
 
So you are going to need something like an Nvidia Geforce 280GTX or an ATI
Radeon HD 4870. You will need at least 512MB of RAM on the card as well.
Also these cards cost in the order of $500-600 US dollars.
 
Now slapping a highend card in a lower end computer is not going to fix
things either. To match the card you will need to have a good CPU (dual coe
is fine, there are very, very few programs coded for more than 2 cores
currently) and the more system RAM the better. If you are playing with large
data sets (eg 2000 images per 3D stack) you will need to go beyond the
maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16
or 32 GB RAM into it.
 
Fast hardrives will always help a little bit but not too much.
 
Also don't worry about going for a two graphics cards running togther
(Nvidia SLI or ATI CrossFire), as far as i know there are no modeling
programs out there yet that can leaverage both cards.
 
Hope that helps you out, if you need any more help just let me know.
 
 
 
Cheers
 
 
Cam
 
 
 
Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
Actually most image display and manipulation software uses your CPU rather
than your GPU.  The CPU does all the number crunching then hands the
resulting data off to the GPU (video card) to display it.  There are a
handful of programs that will actually use the GPU instead, but they are few
and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your
computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]>
wrote:


        Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

        Hi all,

        I would like to know what are the best graphic cards to add in your
PC for microscopic 3D image analysis and manipulation.

        Mine is currently not up to the task.

        Thanks

        Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.  
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
I wasn't aware of that the Olympus and Leica systems actually took advantage of GPUs.  Any thoughts as to what sort of crunching they use them for?
Given the tremendous variety of video cards, how would one go about matching optimal card to optimal software?  At what point do you have to worry about basic CPU speeds as well? (i.e. GPU vs CPU mismatch...?)

Craig


On Wed, Aug 13, 2008 at 8:59 PM, Michael C. Wussow <[hidden email]> wrote:
Hello,

       I have to completely agree with Cam.  His comments are spot on and
what we at Bitplane recommend to our customers.  However, it is recommend
that you check with the company that you are getting the software from just
to make sure there are no incompatibilities between the software and the
graphics card.

If you want specific recommendations don't hesitate to contact us offline.

Mike

Bitplane Inc.
Michael C. Wussow
Vice President and General Manager Bitplane Inc.
 
Cell Phone:    651-336-4600
Fax:                 866-691-9112
Toll Free:       1-888-3D-BITPX (332-4879)
Visit Our Web Site At:  www.bitplane.com



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, August 13, 2008 3:45 PM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
          I owuld like to disagree with Craig. There are now quite a few
programs that use the graphics card to process 3D data in real time now.
MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to
name a few.

There is an enourmous ammount of choice out there in the graphics card
market, but my recomendation (and the recomendation of many other people) is
to buy the best GAMING card you can afford. While each graphics card
manuafacturer has "Workstation" class cards that cost a heap of money, these
cards oare not really any better than a good gaming card. So avoid things
like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.

So you are going to need something like an Nvidia Geforce 280GTX or an ATI
Radeon HD 4870. You will need at least 512MB of RAM on the card as well.
Also these cards cost in the order of $500-600 US dollars.

Now slapping a highend card in a lower end computer is not going to fix
things either. To match the card you will need to have a good CPU (dual coe
is fine, there are very, very few programs coded for more than 2 cores
currently) and the more system RAM the better. If you are playing with large
data sets (eg 2000 images per 3D stack) you will need to go beyond the
maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16
or 32 GB RAM into it.

Fast hardrives will always help a little bit but not too much.

Also don't worry about going for a two graphics cards running togther
(Nvidia SLI or ATI CrossFire), as far as i know there are no modeling
programs out there yet that can leaverage both cards.

Hope that helps you out, if you need any more help just let me know.



Cheers


Cam



Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather
than your GPU.  The CPU does all the number crunching then hands the
resulting data off to the GPU (video card) to display it.  There are a
handful of programs that will actually use the GPU instead, but they are few
and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your
computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]>
wrote:


       Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

       Hi all,

       I would like to know what are the best graphic cards to add in your
PC for microscopic 3D image analysis and manipulation.

       Mine is currently not up to the task.

       Thanks

       Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

The Olympus and Leica software only leverage the graphics card for the 3D modelling part of their software. All other parts rely on the CPU and RAM.

 

While there are a large array of graphics cards out there, they can basically be broken down into two groups Nvida GeForce and ATI Radeon. There are heaps of variations of each of these but in the end they are all built around the same core GPU unit. Some may be clocked slower, have less memory, have slower memory or have parts of the GPU disabled, but they are essentially all the same.

 

Some software will work better on an ATI card vs Nvidia and visa versa, but mostly it will just be a speed difference. You would be pretty unlucky to get a bit of software that would crash constantly on one brand of card.

 

CPU speeds is something that could vary a lot even with the same application. It will depend on what you are doing. If you are just throwing around a pretty 3D model, the GPU will be doing the majority of the work. But if you start doing some analysis of that 3D model you will start to ramp up your CPU usage. The speed of your CPU will only affect how fast a set of data can be crunched, the amount of RAM in your machine will govern how big that data set can be. So if you are happy to wait a little, a low end CPU with lots of RAM would be a better combination.

 

I am sure you can hit a bottleneck (CPU or GPU becoming the limit) with analysis software if you give it enough data. But as long as you have a decent mid to high range CPU, and while i have been an AMD fan boy for years that CPU should be a dual core Intel chip (something from the 9000 series with large cache memory would be good). Couple that with a good GPU board and you will be fine. Oh i guess one other thing is make sure the mother board is using the newer Intel P35 chipset as well.

 

Really it comes down to how much money you have to spend and what you are trying to model/analyse.

 

 

Cheers

 

 

Cam

 

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Thursday, 14 August 2008 1:03 PM
To: [hidden email]
Subject: Re: graphic cards

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I wasn't aware of that the Olympus and Leica systems actually took advantage of GPUs.  Any thoughts as to what sort of crunching they use them for?
Given the tremendous variety of video cards, how would one go about matching optimal card to optimal software?  At what point do you have to worry about basic CPU speeds as well? (i.e. GPU vs CPU mismatch...?)

Craig

On Wed, Aug 13, 2008 at 8:59 PM, Michael C. Wussow <[hidden email]> wrote:

Hello,

       I have to completely agree with Cam.  His comments are spot on and
what we at Bitplane recommend to our customers.  However, it is recommend
that you check with the company that you are getting the software from just
to make sure there are no incompatibilities between the software and the
graphics card.

If you want specific recommendations don't hesitate to contact us offline.

Mike

Bitplane Inc.
Michael C. Wussow
Vice President and General Manager Bitplane Inc.
 
Cell Phone:    651-336-4600
Fax:                 866-691-9112
Toll Free:       1-888-3D-BITPX (332-4879)
Visit Our Web Site At:  www.bitplane.com




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, August 13, 2008 3:45 PM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
          I owuld like to disagree with Craig. There are now quite a few
programs that use the graphics card to process 3D data in real time now.
MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to
name a few.

There is an enourmous ammount of choice out there in the graphics card
market, but my recomendation (and the recomendation of many other people) is
to buy the best GAMING card you can afford. While each graphics card
manuafacturer has "Workstation" class cards that cost a heap of money, these
cards oare not really any better than a good gaming card. So avoid things
like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.

So you are going to need something like an Nvidia Geforce 280GTX or an ATI
Radeon HD 4870. You will need at least 512MB of RAM on the card as well.
Also these cards cost in the order of $500-600 US dollars.

Now slapping a highend card in a lower end computer is not going to fix
things either. To match the card you will need to have a good CPU (dual coe
is fine, there are very, very few programs coded for more than 2 cores
currently) and the more system RAM the better. If you are playing with large
data sets (eg 2000 images per 3D stack) you will need to go beyond the
maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16
or 32 GB RAM into it.

Fast hardrives will always help a little bit but not too much.

Also don't worry about going for a two graphics cards running togther
(Nvidia SLI or ATI CrossFire), as far as i know there are no modeling
programs out there yet that can leaverage both cards.

Hope that helps you out, if you need any more help just let me know.



Cheers


Cam



Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather
than your GPU.  The CPU does all the number crunching then hands the
resulting data off to the GPU (video card) to display it.  There are a
handful of programs that will actually use the GPU instead, but they are few
and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your
computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]>
wrote:


       Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

       Hi all,

       I would like to know what are the best graphic cards to add in your
PC for microscopic 3D image analysis and manipulation.

       Mine is currently not up to the task.

       Thanks

       Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

 

No virus found in this incoming message.
Checked by AVG - http://www.avg.com
Version: 8.0.138 / Virus Database: 270.6.2/1609 - Release Date: 13/08/2008 6:43 AM

This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee. If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email. Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

Engstrom, Lars Engstrom, Lars
Reply | Threaded
Open this post in threaded view
♦
♦
|

dsRED tissue prep

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello all -
A Coworker came to me today looking for a protocol for tissue processing dsRED labeled leukocytes.
What fixatives are compatible with dsRED?
If we are only interested in the dsRED cells do we need to fix the tissue?
Can we simply freeze fresh tissue in OCT and section?
 
Thank you for your time. 
-Lars
Donnelly, Tom Donnelly, Tom
Reply | Threaded
Open this post in threaded view
♦
♦
|

Objective cleaning solution

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

This may be old news to everyone, but A.J. Funk has released an Optical Microscope Lens Cleaner version of Sparkle.  http://www.bestcleaningproducts.com/shop/product.asp?prodID=3779

I am guessing it is the same thing as the purple Sparkle but without the dye that can leave a fluorescent residue.

No commercial interest.

Tom
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
CPU speeds is something that could vary a lot even with the same application. It will depend on what you are doing. If you are just throwing around a pretty 3D model, the GPU will be doing the majority of the work. But if you start doing some analysis of that 3D model you will start to ramp up your CPU usage. The speed of your CPU will only affect how fast a set of data can be crunched, the amount of RAM in your machine will govern how big that data set can be. So if you are happy to wait a little, a low end CPU with lots of RAM would be a better combination.

I was thinking mainly of analysis when I made my earlier comment that the GPU might not be used as much.  So in terms of displaying the data, the GPU is king, but in performing actual statistics or the like on an image it's mainly the CPU that's doing the work, correct?  Are you aware of software that uses the GPU for the actual analysis?

Craig


 

 

I am sure you can hit a bottleneck (CPU or GPU becoming the limit) with analysis software if you give it enough data. But as long as you have a decent mid to high range CPU, and while i have been an AMD fan boy for years that CPU should be a dual core Intel chip (something from the 9000 series with large cache memory would be good). Couple that with a good GPU board and you will be fine. Oh i guess one other thing is make sure the mother board is using the newer Intel P35 chipset as well.

 

Really it comes down to how much money you have to spend and what you are trying to model/analyse.

 

 

Cheers

 

 

Cam

 

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Thursday, 14 August 2008 1:03 PM


To: [hidden email]
Subject: Re: graphic cards

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I wasn't aware of that the Olympus and Leica systems actually took advantage of GPUs.  Any thoughts as to what sort of crunching they use them for?
Given the tremendous variety of video cards, how would one go about matching optimal card to optimal software?  At what point do you have to worry about basic CPU speeds as well? (i.e. GPU vs CPU mismatch...?)

Craig

On Wed, Aug 13, 2008 at 8:59 PM, Michael C. Wussow <[hidden email]> wrote:

Hello,

       I have to completely agree with Cam.  His comments are spot on and
what we at Bitplane recommend to our customers.  However, it is recommend
that you check with the company that you are getting the software from just
to make sure there are no incompatibilities between the software and the
graphics card.

If you want specific recommendations don't hesitate to contact us offline.

Mike

Bitplane Inc.
Michael C. Wussow
Vice President and General Manager Bitplane Inc.
 
Cell Phone:    651-336-4600
Fax:                 866-691-9112
Toll Free:       1-888-3D-BITPX (332-4879)
Visit Our Web Site At:  www.bitplane.com




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, August 13, 2008 3:45 PM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
          I owuld like to disagree with Craig. There are now quite a few
programs that use the graphics card to process 3D data in real time now.
MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to
name a few.

There is an enourmous ammount of choice out there in the graphics card
market, but my recomendation (and the recomendation of many other people) is
to buy the best GAMING card you can afford. While each graphics card
manuafacturer has "Workstation" class cards that cost a heap of money, these
cards oare not really any better than a good gaming card. So avoid things
like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.

So you are going to need something like an Nvidia Geforce 280GTX or an ATI
Radeon HD 4870. You will need at least 512MB of RAM on the card as well.
Also these cards cost in the order of $500-600 US dollars.

Now slapping a highend card in a lower end computer is not going to fix
things either. To match the card you will need to have a good CPU (dual coe
is fine, there are very, very few programs coded for more than 2 cores
currently) and the more system RAM the better. If you are playing with large
data sets (eg 2000 images per 3D stack) you will need to go beyond the
maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16
or 32 GB RAM into it.

Fast hardrives will always help a little bit but not too much.

Also don't worry about going for a two graphics cards running togther
(Nvidia SLI or ATI CrossFire), as far as i know there are no modeling
programs out there yet that can leaverage both cards.

Hope that helps you out, if you need any more help just let me know.



Cheers


Cam



Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather
than your GPU.  The CPU does all the number crunching then hands the
resulting data off to the GPU (video card) to display it.  There are a
handful of programs that will actually use the GPU instead, but they are few
and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your
computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]>
wrote:


       Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

       Hi all,

       I would like to know what are the best graphic cards to add in your
PC for microscopic 3D image analysis and manipulation.

       Mine is currently not up to the task.

       Thanks

       Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

 

No virus found in this incoming message.
Checked by AVG - http://www.avg.com
Version: 8.0.138 / Virus Database: 270.6.2/1609 - Release Date: 13/08/2008 6:43 AM

This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee. If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email. Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.


Aryeh Weiss Aryeh Weiss
Reply | Threaded
Open this post in threaded view
♦
♦
|

question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been told by the Leica rep here that we may put no program on the SP5
computer -- not ImageJ, not Winzip, nothing. We were also told that the computer
may not be connected to the network, except when they want to do service.
To more exact, we were told that if we want anything on the machine, we need to
ask Leica to do it (and to pay for it to be done).


Is this how all  the SP5 users out there are using their systems?

Aside from the obvious  reasons that this policy bothers us, we have a problem
because the IT  people in the university will not permit a computer to be
connected to the network unless they have installed their approved anti-virus
program personally.

If the issue is delicate, perhaps you may prefer to reply off list.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I think there is some confusion here.  I think you are talking about
the low-cost SPE, which runs a really cut-down OS.  I don't think any
of this applies to the SP5.

                                                       Guy


Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Aryeh Weiss
Sent: Friday, 15 August 2008 6:17 AM
To: [hidden email]
Subject: question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been told by the Leica rep here that we may put no program on the SP5 computer -- not ImageJ, not Winzip, nothing. We were also told that the computer may not be connected to the network, except when they want to do service.
To more exact, we were told that if we want anything on the machine, we need to ask Leica to do it (and to pay for it to be done).


Is this how all  the SP5 users out there are using their systems?

Aside from the obvious  reasons that this policy bothers us, we have a problem because the IT  people in the university will not permit a computer to be connected to the network unless they have installed their approved anti-virus program personally.

If the issue is delicate, perhaps you may prefer to reply off list.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date: 13/08/2008 4:14 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date: 13/08/2008 4:14 PM
 
Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: graphic cards

In reply to this post by Craig Brideau
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
You are spot on. The only software i know of that uses a GPU for analysis is the GPU client for the folding at home project. Now this raises a very interesting issue. Folding at home uses the Nvidia CUDA language to run applications on the GPU. The GPU has a much faster floating point unit so therefore can crunch raw data more efficently. For example when running the CPU version of the folding at home client on my dual core 2.8Ghz 12MB cache CPU i get around 100 iterations per second, but when i change over to the GPU version runnign through 2 512MB 8800GTX nvidia cards i get around 2000 iterations per second. So GPU are great for crunching data if the program is written for them. Sadly though there are no imaging apps yet that take advantage of this, in the future i am sure there will be.
 
 
Cheers
 
 
Cam
 
 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Friday, 15 August 2008 5:46 AM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
CPU speeds is something that could vary a lot even with the same application. It will depend on what you are doing. If you are just throwing around a pretty 3D model, the GPU will be doing the majority of the work. But if you start doing some analysis of that 3D model you will start to ramp up your CPU usage. The speed of your CPU will only affect how fast a set of data can be crunched, the amount of RAM in your machine will govern how big that data set can be. So if you are happy to wait a little, a low end CPU with lots of RAM would be a better combination.

I was thinking mainly of analysis when I made my earlier comment that the GPU might not be used as much.  So in terms of displaying the data, the GPU is king, but in performing actual statistics or the like on an image it's mainly the CPU that's doing the work, correct?  Are you aware of software that uses the GPU for the actual analysis?

Craig


 

 

I am sure you can hit a bottleneck (CPU or GPU becoming the limit) with analysis software if you give it enough data. But as long as you have a decent mid to high range CPU, and while i have been an AMD fan boy for years that CPU should be a dual core Intel chip (something from the 9000 series with large cache memory would be good). Couple that with a good GPU board and you will be fine. Oh i guess one other thing is make sure the mother board is using the newer Intel P35 chipset as well.

 

Really it comes down to how much money you have to spend and what you are trying to model/analyse.

 

 

Cheers

 

 

Cam

 

 

 

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Thursday, 14 August 2008 1:03 PM


To: [hidden email]
Subject: Re: graphic cards

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I wasn't aware of that the Olympus and Leica systems actually took advantage of GPUs.  Any thoughts as to what sort of crunching they use them for?
Given the tremendous variety of video cards, how would one go about matching optimal card to optimal software?  At what point do you have to worry about basic CPU speeds as well? (i.e. GPU vs CPU mismatch...?)

Craig

On Wed, Aug 13, 2008 at 8:59 PM, Michael C. Wussow <[hidden email]> wrote:

Hello,

       I have to completely agree with Cam.  His comments are spot on and
what we at Bitplane recommend to our customers.  However, it is recommend
that you check with the company that you are getting the software from just
to make sure there are no incompatibilities between the software and the
graphics card.

If you want specific recommendations don't hesitate to contact us offline.

Mike

Bitplane Inc.
Michael C. Wussow
Vice President and General Manager Bitplane Inc.
 
Cell Phone:    651-336-4600
Fax:                 866-691-9112
Toll Free:       1-888-3D-BITPX (332-4879)
Visit Our Web Site At:  www.bitplane.com




-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nowell, Cameron
Sent: Wednesday, August 13, 2008 3:45 PM
To: [hidden email]
Subject: Re: graphic cards

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi List,
          I owuld like to disagree with Craig. There are now quite a few
programs that use the graphics card to process 3D data in real time now.
MetaMorph, Imaris, Volocity, 3D Studio, Olympus FV1000, Leica LAS AF just to
name a few.

There is an enourmous ammount of choice out there in the graphics card
market, but my recomendation (and the recomendation of many other people) is
to buy the best GAMING card you can afford. While each graphics card
manuafacturer has "Workstation" class cards that cost a heap of money, these
cards oare not really any better than a good gaming card. So avoid things
like the Nvidia Quadro, ATI FirePro or Wildcat Studio type of cards.

So you are going to need something like an Nvidia Geforce 280GTX or an ATI
Radeon HD 4870. You will need at least 512MB of RAM on the card as well.
Also these cards cost in the order of $500-600 US dollars.

Now slapping a highend card in a lower end computer is not going to fix
things either. To match the card you will need to have a good CPU (dual coe
is fine, there are very, very few programs coded for more than 2 cores
currently) and the more system RAM the better. If you are playing with large
data sets (eg 2000 images per 3D stack) you will need to go beyond the
maximum 4GB of RAM in a 32 bit system and go for a 64Bit system and put 16
or 32 GB RAM into it.

Fast hardrives will always help a little bit but not too much.

Also don't worry about going for a two graphics cards running togther
(Nvidia SLI or ATI CrossFire), as far as i know there are no modeling
programs out there yet that can leaverage both cards.

Hope that helps you out, if you need any more help just let me know.



Cheers


Cam



Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia

________________________________

From: Confocal Microscopy List on behalf of Craig Brideau
Sent: Thu 14/08/2008 6:16 AM
To: [hidden email]
Subject: Re: graphic cards


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Actually most image display and manipulation software uses your CPU rather
than your GPU.  The CPU does all the number crunching then hands the
resulting data off to the GPU (video card) to display it.  There are a
handful of programs that will actually use the GPU instead, but they are few
and far between as they require special programming.
What sort of processor, video card, and how much RAM do you have in your
computer?

Craig



On Wed, Aug 13, 2008 at 2:04 PM, Marc Thibault <[hidden email]>
wrote:


       Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

       Hi all,

       I would like to know what are the best graphic cards to add in your
PC for microscopic 3D image analysis and manipulation.

       Mine is currently not up to the task.

       Thanks

       Marc



This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

 

No virus found in this incoming message.
Checked by AVG - http://www.avg.com
Version: 8.0.138 / Virus Database: 270.6.2/1609 - Release Date: 13/08/2008 6:43 AM

This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee. If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email. Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.


This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee. If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email. Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

Larry Tague Larry Tague
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective cleaning solution

In reply to this post by Donnelly, Tom
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Yea, I saw the Funk new microscope objective cleaner, but we have been
using the purple stuff for a long time on Zeiss and Nikon objectives
with no problems.... of course I do dilute it 1 part Sparkle to 3 parts
distilled water.  I understand if you don't like the color, you can
leave it in the sun for a day or so and it turns clear.  Never tried
that... not needed in my opinion.

Larry

Donnelly, Tom wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> This may be old news to everyone, but A.J. Funk has released an Optical Microscope Lens Cleaner version of Sparkle.  http://www.bestcleaningproducts.com/shop/product.asp?prodID=3779
>
> I am guessing it is the same thing as the purple Sparkle but without the dye that can leave a fluorescent residue.
>
> No commercial interest.
>
> Tom
>  

--
Larry Tague
Co-Investigator BBHSL*
Co-Director of MECCA**
Research Associate (Dept. of Physiology)
University of Tennessee Health Science Center
894 Union Ave.
Memphis, TN 38163
Phone Bus.: 901-448-7152
Phone FAX:  901-448-7126
e-mail:[hidden email] or
       [hidden email]

*BBHSL "Building Bridges to Health Science Literacy" URL: http://bbhsl.mecca.org, a Science Education Partnership Award (SEPA). Supported by the National Center of Research Resources (NCRR) of the National Institutes of Health (NIH).
**MECCA (Memphis Educational Computer Connectivity Alliance)
URL: http://www.mecca.org/.  Originally support by the National Science Foundation's "Networking Infrastructure or Education" program.

CONFIDENTIALITY NOTICE: This email and any files transmitted with it are confidential and are intended solely for the use of the individual or entity to whom they are addressed. This communication may contain protected health information, or other legally privileged, confidential, or proprietary information. If you are not the intended recipient or the individual responsible for delivering the email to the recipient, please be advised that you have received this email in error and that any use, dissemination, forwarding, printing, or copying of this email is strictly prohibited. If you have received this email in error, please immediately notify the sender, disregard the foregoing message, and delete the message. We apologize for any inconvenience this may have caused. IMPORTANT NOTE: Confidential health   information is protected by state and federal law, including, but not limited to, the Health Insurance Portability and Accountability Act of 1996 and related regulations.
Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question for Leica SP5 users

In reply to this post by Guy Cox
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have an SP5 in our facility and the only thing we have been told by
leica we can't do is connect it to a network, which is fine for us as
the network here is notoriously unstable.

We run our own anti-virus software and defrag utilities on it, as well
as a few admin apps to track data usage etc.


Cheers


Cam


Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Guy Cox
Sent: Friday, 15 August 2008 8:10 AM
To: [hidden email]
Subject: Re: question for Leica SP5 users


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I think there is some confusion here.  I think you are talking about the
low-cost SPE, which runs a really cut-down OS.  I don't think any of
this applies to the SP5.

                                                       Guy


Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit,
Madsen Building F09, University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Aryeh Weiss
Sent: Friday, 15 August 2008 6:17 AM
To: [hidden email]
Subject: question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been told by the Leica rep here that we may put no program on
the SP5 computer -- not ImageJ, not Winzip, nothing. We were also told
that the computer may not be connected to the network, except when they
want to do service.
To more exact, we were told that if we want anything on the machine, we
need to ask Leica to do it (and to pay for it to be done).


Is this how all  the SP5 users out there are using their systems?

Aside from the obvious  reasons that this policy bothers us, we have a
problem because the IT  people in the university will not permit a
computer to be connected to the network unless they have installed their
approved anti-virus program personally.

If the issue is delicate, perhaps you may prefer to reply off list.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
13/08/2008 4:14 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
13/08/2008 4:14 PM
 

This email (including any attachments or links) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee.  If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email.  Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question for Leica SP5 users

In reply to this post by Aryeh Weiss
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I haven't checked with Leica but I find this hard to believe.
There's no way we could use any system that could not be networked
(since we run a central server for all digital images) and I
would reckon that the same would apply for more than 90% of all
facilities that would buy a high-end confocal.  Maybe Leica can
reply to this?

                                                  Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Nowell, Cameron
Sent: Friday, 15 August 2008 8:30 AM
To: [hidden email]
Subject: Re: question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have an SP5 in our facility and the only thing we have been told by leica we can't do is connect it to a network, which is fine for us as the network here is notoriously unstable.

We run our own anti-virus software and defrag utilities on it, as well as a few admin apps to track data usage etc.


Cheers


Cam


Cameron Nowell
Research and Microscopy Imaging Core
Peter MacCallum Cancer Centre
Melbourne, Australia



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox
Sent: Friday, 15 August 2008 8:10 AM
To: [hidden email]
Subject: Re: question for Leica SP5 users


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I think there is some confusion here.  I think you are talking about the low-cost SPE, which runs a really cut-down OS.  I don't think any of this applies to the SP5.

                                                       Guy


Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Aryeh Weiss
Sent: Friday, 15 August 2008 6:17 AM
To: [hidden email]
Subject: question for Leica SP5 users

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have been told by the Leica rep here that we may put no program on the SP5 computer -- not ImageJ, not Winzip, nothing. We were also told that the computer may not be connected to the network, except when they want to do service.
To more exact, we were told that if we want anything on the machine, we need to ask Leica to do it (and to pay for it to be done).


Is this how all  the SP5 users out there are using their systems?

Aside from the obvious  reasons that this policy bothers us, we have a problem because the IT  people in the university will not permit a computer to be connected to the network unless they have installed their approved anti-virus program personally.

If the issue is delicate, perhaps you may prefer to reply off list.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
13/08/2008 4:14 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
13/08/2008 4:14 PM
 

This email (including any attachments or links) may contain confidential and/or legally privileged information and is intended only to be read or used by the addressee.  If you are not the intended addressee, any use, distribution, disclosure or copying of this email is strictly prohibited.  
Confidentiality and legal privilege attached to this email (including any attachments) are not waived or lost by reason of its mistaken delivery to you.
If you have received this email in error, please delete it and notify us immediately by telephone or email.  Peter MacCallum Cancer Centre provides no guarantee that this transmission is free of virus or that it has not been intercepted or altered and will not be liable for any delay in its receipt.

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date: 13/08/2008 4:14 PM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date: 13/08/2008 4:14 PM
 
Lenaldo Rocha Lenaldo Rocha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question for Leica SP5 users

In reply to this post by Nowell, Cameron
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Our SP5 is connected to our network. I even installed a vnc server on  
it and used to live broadcast over the internet the images captured  
during a course.

Lenaldo


On 14/08/2008, at 19:29, Nowell, Cameron wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have an SP5 in our facility and the only thing we have been told by
> leica we can't do is connect it to a network, which is fine for us as
> the network here is notoriously unstable.
>
> We run our own anti-virus software and defrag utilities on it, as well
> as a few admin apps to track data usage etc.
>
>
> Cheers
>
>
> Cam
>
>
> Cameron Nowell
> Research and Microscopy Imaging Core
> Peter MacCallum Cancer Centre
> Melbourne, Australia
>
>
>
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On
> Behalf Of Guy Cox
> Sent: Friday, 15 August 2008 8:10 AM
> To: [hidden email]
> Subject: Re: question for Leica SP5 users
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I think there is some confusion here.  I think you are talking about  
> the
> low-cost SPE, which runs a really cut-down OS.  I don't think any of
> this applies to the SP5.
>
>                                                       Guy
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>    http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit,
> Madsen Building F09, University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>     http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On
> Behalf Of Aryeh Weiss
> Sent: Friday, 15 August 2008 6:17 AM
> To: [hidden email]
> Subject: question for Leica SP5 users
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have been told by the Leica rep here that we may put no program on
> the SP5 computer -- not ImageJ, not Winzip, nothing. We were also told
> that the computer may not be connected to the network, except when  
> they
> want to do service.
> To more exact, we were told that if we want anything on the machine,  
> we
> need to ask Leica to do it (and to pay for it to be done).
>
>
> Is this how all  the SP5 users out there are using their systems?
>
> Aside from the obvious  reasons that this policy bothers us, we have a
> problem because the IT  people in the university will not permit a
> computer to be connected to the network unless they have installed  
> their
> approved anti-virus program personally.
>
> If the issue is delicate, perhaps you may prefer to reply off list.
>
> --aryeh
> --
> Aryeh Weiss
> School of Engineering
> Bar Ilan University
> Ramat Gan 52900 Israel
>
> Ph:  972-3-5317638
> FAX: 972-3-7384050
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
> 13/08/2008 4:14 PM
>
>
> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.526 / Virus Database: 270.6.3/1610 - Release Date:
> 13/08/2008 4:14 PM
>
>
> This email (including any attachments or links) may contain
> confidential and/or legally privileged information and is
> intended only to be read or used by the addressee.  If you
> are not the intended addressee, any use, distribution,
> disclosure or copying of this email is strictly
> prohibited.
> Confidentiality and legal privilege attached to this email
> (including any attachments) are not waived or lost by
> reason of its mistaken delivery to you.
> If you have received this email in error, please delete it
> and notify us immediately by telephone or email.  Peter
> MacCallum Cancer Centre provides no guarantee that this
> transmission is free of virus or that it has not been
> intercepted or altered and will not be liable for any delay
> in its receipt.
>
Aryeh Weiss Aryeh Weiss
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question for Leica SP5 users - followup

In reply to this post by Guy Cox
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have received many replies to this -- most of them off list. So first, thank
you to the many  Leica users who replied. There are too many  (and more coming
in, to reply personally to everyone who wrote, but it is greatly appreciated.

The consensus of the replies (so far about a dozen replies) is that one must
exercise care, but most everyone who wrote has added something to their system
(usually ImageJ, admin utilities, and similar). Most have their computer
networked, with some antivirus. None appear to have been ordered by Leica not to
  put anything on their machine, but Leica did recommend (at least to some)that
other programs not be added.

One responder asked why we would want to add anything. We of course do not
intend to turn this into a general use computer, but there are some additions
that I consider important.  I should add that we have not yet received the
machine -- the rep had visited to inspect the installation site, and apparently
he wanted to "lay down the law". So I am not sure exactly what it included in
the Leica software installation.

1. A compression utility, such as WINZIP. Last time I saw the Leica image
format, it was uncompressed. Lossless LZW compression can save 30-50%, depending
on the images, and even more if you have two channels stored in uncompressed RGB
(becasue the third channel is identically zero).  The datasets can be very
large, and transferring compressed files can save time.

2. ImageJ + bioformats plugin. We prefer this to the utilities that the
manufacturers sometimes provide to read their special formats. We often put
ImageJ+appropriate plugins onto data DVDs  that we burn. It saves time because
the users have something with which they can read their files. We can tell them
to install ImageJ at their end until we are blue in the face -- it is easier to
just give them a working copy.

3. Network -- I prefer to leave a machine which is dedicated to running hardware
off of the network. However, if Leica is anyway going to make us connect it to
the network, then I want to use that to transfer intermediate sized data sets
(too small to justify burning a DVD, too large for a flashdisk). I know DVDs are
cheap, but there are ecological considerations... BTW -- a few responses
mentioned that they are running antivirus software without problems.

4. Someone warned against installing Microscoft office on the system. I would
extend this warning any Microsoft software. They apparently believe that it is
ok for software to fiddle with system files,  drivers, whatever. However,
OpenOffice would be nice, mainly for being able to conveniently make notes about
the system or the experiment.

5. Finally, we may want to put some tracking or auditing software on the system.

Someone asked if we were talking about an SPE, because that system has real
restrictions on its OS. I can assure you all that are talking about an SP5.

--aryeh
--
Aryeh Weiss
School of Engineering
Bar Ilan University
Ramat Gan 52900 Israel

Ph:  972-3-5317638
FAX: 972-3-7384050
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective cleaning solution

In reply to this post by Larry Tague
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Sparkle does indeed bleach if left in the sun. I discovered this
accidentally, and the bleaching of the color has no effect on
Sparkle's cleaning ability.
As near as I can tell, the new lens cleaner version is just
non-purple Sparkle at ~13X the price ($30/0.75 L vs $3/1 L). Guess
they're aiming at the hospital market.

I've used Sparkly clear and purple for years now on fluorescence
microscopes, and haven't noticed any residue problems.

Phil

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Yea, I saw the Funk new microscope objective cleaner, but we have
>been using the purple stuff for a long time on Zeiss and Nikon
>objectives with no problems.... of course I do dilute it 1 part
>Sparkle to 3 parts distilled water.  I understand if you don't like
>the color, you can leave it in the sun for a day or so and it turns
>clear.  Never tried that... not needed in my opinion.
>
>Larry
>
>Donnelly, Tom wrote:
>>  Search the CONFOCAL archive at
>>  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>  This may be old news to everyone, but A.J. Funk has released an
>>Optical Microscope Lens Cleaner version of Sparkle.
>>http://www.bestcleaningproducts.com/shop/product.asp?prodID=3779
>>
>>  I am guessing it is the same thing as the purple Sparkle but
>>without the dye that can leave a fluorescent residue.
>>  No commercial interest.
>>  Tom
>>
>
>--
>Larry Tague
>Co-Investigator BBHSL*
>Co-Director of MECCA**
>Research Associate (Dept. of Physiology)
>University of Tennessee Health Science Center
>894 Union Ave.
>Memphis, TN 38163
>Phone Bus.: 901-448-7152
>Phone FAX:  901-448-7126
>e-mail:[hidden email] or
>       [hidden email]
--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576