Hi there! Does anyone know of a green nuclear stain
that resists fixation? One of our users wants to do some colocalization of cy3
and cy5 with the nucleus and our confocal has unfortunately no 405 laser. We
are trying to avoid NLS-GFP but use something that would label all cells right
before or after fixation. Thanks to all! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Dept of Biosciences and Nutrition Karolinska Institutet Novum 14157 Huddinge Sweden +46 (0)8 608 9240 |
![]() |
Hi Sylvie,
I have had great success in the past using OliGreen from Invitrogen (no commercial affiliation), though it is not supplied specifically as a DNA stain. Cheers Cam Cameron J. Nowell Microscpy Manager Central Resource for Advanced Microscopy Ludwig Insttue for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 http://www.ludwig.edu.au/branch/research/platform/microscopy.htm ________________________________ From: Confocal Microscopy List on behalf of Sylvie Le Guyader Sent: Wed 30/09/2009 7:20 PM To: [hidden email] Subject: green nuclear stain Hi there! Does anyone know of a green nuclear stain that resists fixation? One of our users wants to do some colocalization of cy3 and cy5 with the nucleus and our confocal has unfortunately no 405 laser. We are trying to avoid NLS-GFP but use something that would label all cells right before or after fixation. Thanks to all! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Dept of Biosciences and Nutrition Karolinska Institutet Novum 14157 Huddinge Sweden +46 (0)8 608 9240 |
![]() |
In reply to this post by Sylvie Le Guyader-2
Yopro from Invitrogen is a green fluorescent nuclear counterstain.
From: Sylvie Le Guyader Date: Wed, 30 Sep 2009 11:20:08 +0200 To: <[hidden email]> Subject: green nuclear stain Hi there! Does anyone know of a green nuclear stain that resists fixation? One of our users wants to do some colocalization of cy3 and cy5 with the nucleus and our confocal has unfortunately no 405 laser. We are trying to avoid NLS-GFP but use something that would label all cells right before or after fixation. Thanks to all! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Dept of Biosciences and Nutrition Karolinska Institutet Novum 14157 Huddinge Sweden +46 (0)8 608 9240 |
![]() |
In reply to this post by Sylvie Le Guyader-2
Tjänare Sylvie,
you might try Sytox Green or TO-PRO 1. Michael > Hi there! > > > > Does anyone know of a green nuclear stain that resists fixation? One of > our > users wants to do some colocalization of cy3 and cy5 with the nucleus and > our confocal has unfortunately no 405 laser. We are trying to avoid > NLS-GFP > but use something that would label all cells right before or after > fixation. > > > > Thanks to all! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Novum > > 14157 Huddinge > > Sweden > > +46 (0)8 608 9240 |
![]() |
In reply to this post by Sylvie Le Guyader-2
Forgot to say... for colocalization with Cy3 I would prefer the Sytox
Green, unless you are using a 561 nm laser for Cy3. The TO-PRO 1 is still excitable with 543 nm and will therefore bleed into Cy3 which for sure messes up the analysis. Michael > Hi there! > > > > Does anyone know of a green nuclear stain that resists fixation? One of > our > users wants to do some colocalization of cy3 and cy5 with the nucleus and > our confocal has unfortunately no 405 laser. We are trying to avoid > NLS-GFP > but use something that would label all cells right before or after > fixation. > > > > Thanks to all! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Novum > > 14157 Huddinge > > Sweden > > +46 (0)8 608 9240 |
![]() |
In reply to this post by Sylvie Le Guyader-2
Sytox Green (Molecular Probes) works well for fixed, not sure about live
cells. Normal dilution is 1:20,000 but titrate because it is very bright and depending on your setup and other dyes may be seen in other channels unless it is at max dilution. Judy Judy Trogadis Bio-Imaging Coordinator St. Michael's Hospital, 7Queen 30 Bond St. Toronto, ON M5B 1W8, Canada ph: 416-864-6060 x6337 pager: 416-685-9219 fax: 416-864-5046 [hidden email] >>> Sylvie Le Guyader <[hidden email]> 9/30/2009 5:20 AM >>> Hi there! Does anyone know of a green nuclear stain that resists fixation? One of our users wants to do some colocalization of cy3 and cy5 with the nucleus and our confocal has unfortunately no 405 laser. We are trying to avoid NLS-GFP but use something that would label all cells right before or after fixation. Thanks to all! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Dept of Biosciences and Nutrition Karolinska Institutet Novum 14157 Huddinge Sweden +46 (0)8 608 9240 |
![]() |
Dear all
Thanks a lot for all the suggestions. We will try and hopefully get nice results! :) Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Dept of Biosciences and Nutrition Karolinska Institutet Novum 14157 Huddinge Sweden +46 (0)8 608 9240 > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Judy > Trogadis > Sent: 30 September 2009 15:26 > To: [hidden email] > Subject: Re: green nuclear stain > > Sytox Green (Molecular Probes) works well for fixed, not sure about live > cells. Normal dilution is 1:20,000 > but titrate because it is very bright and depending on your setup and > other dyes may be seen in other channels unless it is at max dilution. > > Judy > > Judy Trogadis > Bio-Imaging Coordinator > St. Michael's Hospital, 7Queen > 30 Bond St. > Toronto, ON M5B 1W8, Canada > ph: 416-864-6060 x6337 > pager: 416-685-9219 > fax: 416-864-5046 > [hidden email] > > > >>> Sylvie Le Guyader <[hidden email]> 9/30/2009 5:20 AM >>> > Hi there! > > > > Does anyone know of a green nuclear stain that resists fixation? One of > our > users wants to do some colocalization of cy3 and cy5 with the nucleus > and > our confocal has unfortunately no 405 laser. We are trying to avoid > NLS-GFP > but use something that would label all cells right before or after > fixation. > > > > Thanks to all! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Novum > > 14157 Huddinge > > Sweden > > +46 (0)8 608 9240 > > |
![]() |
Sylvie,
you also might want to have a look at Chromomycin A3. disadvantage: needs Mg-Ions or will not stain (you surely can google some protocols) advantages: DNA-specific (most others also stain RNA); old and patent free and thus comparably cheap (if I remember right). Steffen At 15:40 30.09.2009, you wrote: >Dear all > >Thanks a lot for all the suggestions. We will try and hopefully get nice >results! > >:) > >Med vänlig hälsning / Best regards > >Sylvie > >@@@@@@@@@@@@@@@@@@@@@@@@ >Sylvie Le Guyader >Dept of Biosciences and Nutrition >Karolinska Institutet >Novum >14157 Huddinge >Sweden >+46 (0)8 608 9240 > > > -----Original Message----- > > From: Confocal Microscopy List > > [mailto:[hidden email]] On Behalf Of Judy > > Trogadis > > Sent: 30 September 2009 15:26 > > To: [hidden email] > > Subject: Re: green nuclear stain > > > > Sytox Green (Molecular Probes) works well for fixed, not sure about live > > cells. Normal dilution is 1:20,000 > > but titrate because it is very bright and depending on your setup and > > other dyes may be seen in other channels unless it is at max dilution. > > > > Judy > > > > Judy Trogadis > > Bio-Imaging Coordinator > > St. Michael's Hospital, 7Queen > > 30 Bond St. > > Toronto, ON M5B 1W8, Canada > > ph: 416-864-6060 x6337 > > pager: 416-685-9219 > > fax: 416-864-5046 > > [hidden email] > > > > > > >>> Sylvie Le Guyader <[hidden email]> 9/30/2009 5:20 AM >>> > > Hi there! > > > > > > > > Does anyone know of a green nuclear stain that resists fixation? One of > > our > > users wants to do some colocalization of cy3 and cy5 with the nucleus > > and > > our confocal has unfortunately no 405 laser. We are trying to avoid > > NLS-GFP > > but use something that would label all cells right before or after > > fixation. > > > > > > > > Thanks to all! > > > > > > > > Med vänlig hälsning / Best regards > > > > > > > > Sylvie > > > > > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > > > Sylvie Le Guyader > > > > Dept of Biosciences and Nutrition > > > > Karolinska Institutet > > > > Novum > > > > 14157 Huddinge > > > > Sweden > > > > +46 (0)8 608 9240 > > > > |
![]() |
HI all,
While i haven't used the dye Steffen recommends, i have used DNA dyes that light up RNA as well. Give a quick treatment of your samples with RNase prior to staining will get rid of the problem. Cheers Cam Cameron J. Nowell Microscopy Manager Centre for Advanced Microscopy Ludwig Institute for Cancer Research PO Box 2008 Royal Melbourne Hospital Victoria, 3050 AUSTRALIA Office: +61 3 9341 3155 Mobile: +61422882700 Fax: +61 3 9341 3104 Facility Website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Dietzel Sent: Thursday, 1 October 2009 11:05 PM To: [hidden email] Subject: Re: green nuclear stain Sylvie, you also might want to have a look at Chromomycin A3. disadvantage: needs Mg-Ions or will not stain (you surely can google some protocols) advantages: DNA-specific (most others also stain RNA); old and patent free and thus comparably cheap (if I remember right). Steffen At 15:40 30.09.2009, you wrote: >Dear all > >Thanks a lot for all the suggestions. We will try and hopefully get nice >results! > >:) > >Med vänlig hälsning / Best regards > >Sylvie > >@@@@@@@@@@@@@@@@@@@@@@@@ >Sylvie Le Guyader >Dept of Biosciences and Nutrition >Karolinska Institutet >Novum >14157 Huddinge >Sweden >+46 (0)8 608 9240 > > > -----Original Message----- > > From: Confocal Microscopy List > > [mailto:[hidden email]] On Behalf Of Judy > > Trogadis > > Sent: 30 September 2009 15:26 > > To: [hidden email] > > Subject: Re: green nuclear stain > > > > Sytox Green (Molecular Probes) works well for fixed, not sure about live > > cells. Normal dilution is 1:20,000 > > but titrate because it is very bright and depending on your setup and > > other dyes may be seen in other channels unless it is at max dilution. > > > > Judy > > > > Judy Trogadis > > Bio-Imaging Coordinator > > St. Michael's Hospital, 7Queen > > 30 Bond St. > > Toronto, ON M5B 1W8, Canada > > ph: 416-864-6060 x6337 > > pager: 416-685-9219 > > fax: 416-864-5046 > > [hidden email] > > > > > > >>> Sylvie Le Guyader <[hidden email]> 9/30/2009 5:20 AM >>> > > Hi there! > > > > > > > > Does anyone know of a green nuclear stain that resists fixation? One of > > our > > users wants to do some colocalization of cy3 and cy5 with the nucleus > > and > > our confocal has unfortunately no 405 laser. We are trying to avoid > > NLS-GFP > > but use something that would label all cells right before or after > > fixation. > > > > > > > > Thanks to all! > > > > > > > > Med vänlig hälsning / Best regards > > > > > > > > Sylvie > > > > > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > > > Sylvie Le Guyader > > > > Dept of Biosciences and Nutrition > > > > Karolinska Institutet > > > > Novum > > > > 14157 Huddinge > > > > Sweden > > > > +46 (0)8 608 9240 > > > > No virus found in this incoming message. Checked by AVG - www.avg.com Version: 8.5.409 / Virus Database: 270.13.115/2405 - Release Date: 09/30/09 10:35:00 This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research Ltd does not waiver any rights if you have received this communication in error. The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research Ltd. |