Does anyone have suggestions for imaging hair?  We have someone who wants to get a cross section.  Is there some fluorescent label?  I am having difficulty with the density of the hair (light doesn't penetrate very far).

Elaine Kunze
Cytometry Facility
Huck Institute of the Life Sciences
319 Life Sciences Building
Penn State University
University Park, PA 16802
http://www.huck.psu.edu/facilities/cytometry-up/
814-863-2762

Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

Hi Elaine,

A simple way to get hair cross sections is to pull some of the hair and some
cotton fibres through a small hole in a metal plate, then cut across with a
sharp razor blade on either side of the plate (which shouldn¹t be too
thick...).  Place on a slide, add immersion oil or water, coverslip and
you¹ll see a good cross-seciton of the hairs.  Your workshop should be able
to drill some small holes (can't remember dimensions, but small...) in a
small piece of metal sheeting.

This was part of a zoology prac. a long time ago, comparing hair from
different animals, including members of the class.

cheers,
Rosmary

Dr Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph 61 2 6246 5475
fx 61 2 6246 5334



On 4/12/08 8:27 AM, "Elaine Kunze" <[hidden email]> wrote:

> Does anyone have suggestions for imaging hair?  We have someone who wants to
> get a cross section.  Is there some fluorescent label?  I am having difficulty
> with the density of the hair (light doesn't penetrate very far).
>
> Elaine Kunze
> Cytometry Facility
> Huck Institute of the Life Sciences
> 319 Life Sciences Building
> Penn State University
> University Park, PA 16802
> http://www.huck.psu.edu/facilities/cytometry-up/
>  <http://www.huck.psu.edu/facilities/cytometry-up/> 814-863-2762
Larson Jeffrey M. Larson Jeffrey M.
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

Hi Elaine,

While not a confocal technique, I have gotten beautiful images of hair in transmitted light DIC.  The quality of the image doesn't deteriorate too badly as you focus through and the pigment granules are beautiful because they're highly birefringent.  I've used objective lenses with magnifications as high as 100X for this.  Your students may want to compare different hair colors against their bleached and/or dyed equivalents.  It's also interesting to look at the differences between the brown and white hairs of familiar animals.  It will be difficult to look at hair in cross section with transmitted light DIC unless you can cut very thin sections.

Jeffrey Larson
Product Manager
Nikon Confocal Systems
Nikon Instruments Inc.
(631) 547-8540

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Rosemary White
Sent: Wednesday, December 03, 2008 4:52 PM
To: [hidden email]
Subject: Re: hair

Hi Elaine,

A simple way to get hair cross sections is to pull some of the hair and some
cotton fibres through a small hole in a metal plate, then cut across with a
sharp razor blade on either side of the plate (which shouldn¹t be too
thick...).  Place on a slide, add immersion oil or water, coverslip and
you¹ll see a good cross-seciton of the hairs.  Your workshop should be able
to drill some small holes (can't remember dimensions, but small...) in a
small piece of metal sheeting.

This was part of a zoology prac. a long time ago, comparing hair from
different animals, including members of the class.

cheers,
Rosmary

Dr Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph 61 2 6246 5475
fx 61 2 6246 5334



On 4/12/08 8:27 AM, "Elaine Kunze" <[hidden email]> wrote:

> Does anyone have suggestions for imaging hair?  We have someone who wants to
> get a cross section.  Is there some fluorescent label?  I am having difficulty
> with the density of the hair (light doesn't penetrate very far).
>
> Elaine Kunze
> Cytometry Facility
> Huck Institute of the Life Sciences
> 319 Life Sciences Building
> Penn State University
> University Park, PA 16802
> http://www.huck.psu.edu/facilities/cytometry-up/
>  <http://www.huck.psu.edu/facilities/cytometry-up/> 814-863-2762
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

In reply to this post by Elaine Kunze
Dear Elaine,
vital stains will work in the region of the hair bulb where you have  
some live cells. Incubate the freshly plucked hair with labels diluted  
in culture medium, rinse, then fix with paraformaldehyde.  Some  
general protein stain such as eosin might be helpful in the acellular  
shaft.  Your bigger problem will be the spherical aberration arising  
from refractive index of the hair.  You may need to try different  
clearing agents such as glycerol dilutions, immersion oil or even  
higher RI reagents.

Regards,
Glen

On Dec 3, 2008, at 1:27 PM, Elaine Kunze wrote:

> Does anyone have suggestions for imaging hair?  We have someone who  
> wants to get a cross section.  Is there some fluorescent label?  I  
> am having difficulty with the density of the hair (light doesn't  
> penetrate very far).
> Elaine Kunze
> Cytometry Facility
> Huck Institute of the Life Sciences
> 319 Life Sciences Building
> Penn State University
> University Park, PA 16802
> http://www.huck.psu.edu/facilities/cytometry-up/
> 814-863-2762
>



Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

******************************************************************************
The box said "Requires WindowsXP or better", so I bought a Macintosh.
******************************************************************************
Zoltan Zoltan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

Hi Elaine,

TDE treatment (ending in 96% TDE) seems to work well with hair follicles; I imaged these with 20x oil and 40x oil on a Leica SP5 and was able to see through the bulb.  The hair shafts that I used were  autofluorescent at 561 nm excitation with a rather spiky emission spectrum (anyone knows where this arises from?), giving great images even w/o staining.  I can confirm that DIC Z stacks look great as well, I used the 561 and/or the 488 lines to record this on the SP5.   A bit of deconvolution with theoretical PSF improves the DIC stacks even further. 
  Best,

Zoltan


On Wed, Dec 3, 2008 at 10:54 PM, Glen MacDonald <[hidden email]> wrote:
Dear Elaine,
vital stains will work in the region of the hair bulb where you have some live cells. Incubate the freshly plucked hair with labels diluted in culture medium, rinse, then fix with paraformaldehyde.  Some general protein stain such as eosin might be helpful in the acellular shaft.  Your bigger problem will be the spherical aberration arising from refractive index of the hair.  You may need to try different clearing agents such as glycerol dilutions, immersion oil or even higher RI reagents.

Regards,
Glen


On Dec 3, 2008, at 1:27 PM, Elaine Kunze wrote:

Does anyone have suggestions for imaging hair?  We have someone who wants to get a cross section.  Is there some fluorescent label?  I am having difficulty with the density of the hair (light doesn't penetrate very far).
Elaine Kunze
Cytometry Facility
Huck Institute of the Life Sciences
319 Life Sciences Building
Penn State University
University Park, PA 16802
http://www.huck.psu.edu/facilities/cytometry-up/
814-863-2762




Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

******************************************************************************
The box said "Requires WindowsXP or better", so I bought a Macintosh.
******************************************************************************



--

Zoltan Cseresnyes
Facility manager, Imaging Suite
University of Cambridge, UK
Oshel, Philip Eugene Oshel, Philip Eugene
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

In reply to this post by Larson Jeffrey M.
I agree that DIC can work well for hair imaging, and it's an
excellent complementary technique for SEM of hair. There are some
examples on the Michigan Mammals hair ID pages:
http://www.cst.cmich.edu/users/swans1bj/Hair%20ID/Species.html

Phil

>Hi Elaine,
>
>While not a confocal technique, I have gotten beautiful images of
>hair in transmitted light DIC.  The quality of the image doesn't
>deteriorate too badly as you focus through and the pigment granules
>are beautiful because they're highly birefringent.  I've used
>objective lenses with magnifications as high as 100X for this.  Your
>students may want to compare different hair colors against their
>bleached and/or dyed equivalents.  It's also interesting to look at
>the differences between the brown and white hairs of familiar
>animals.  It will be difficult to look at hair in cross section with
>transmitted light DIC unless you can cut very thin sections.
>
>Jeffrey Larson
>Product Manager
>Nikon Confocal Systems
>Nikon Instruments Inc.
>(631) 547-8540
>
>-----Original Message-----
>From: Confocal Microscopy List
>[mailto:[hidden email]] On Behalf Of Rosemary White
>Sent: Wednesday, December 03, 2008 4:52 PM
>To: [hidden email]
>Subject: Re: hair
>
>Hi Elaine,
>
>A simple way to get hair cross sections is to pull some of the hair and some
>cotton fibres through a small hole in a metal plate, then cut across with a
>sharp razor blade on either side of the plate (which shouldn't be too
>thick...).  Place on a slide, add immersion oil or water, coverslip and
>you'll see a good cross-seciton of the hairs.  Your workshop should be able
>to drill some small holes (can't remember dimensions, but small...) in a
>small piece of metal sheeting.
>
>This was part of a zoology prac. a long time ago, comparing hair from
>different animals, including members of the class.
>
>cheers,
>Rosmary
>
>Dr Rosemary White
>CSIRO Plant Industry
>GPO Box 1600
>Canberra, ACT 2601
>Australia
>
>ph 61 2 6246 5475
>fx 61 2 6246 5334
>
>
>
>On 4/12/08 8:27 AM, "Elaine Kunze" <[hidden email]> wrote:
>
>>  Does anyone have suggestions for imaging hair?  We have someone who wants to
>>  get a cross section.  Is there some fluorescent label?  I am
>>having difficulty
>>  with the density of the hair (light doesn't penetrate very far).
>>
>>  Elaine Kunze
>>  Cytometry Facility
>>  Huck Institute of the Life Sciences
>>  319 Life Sciences Building
>>  Penn State University
>>  University Park, PA 16802
>>  http://www.huck.psu.edu/facilities/cytometry-up/
>>   <http://www.huck.psu.edu/facilities/cytometry-up/> 814-863-2762

--
Philip Oshel
Microscopy Facility Supervisor
Biology Department
024C Brooks Hall
Central Michigan University
Mt. Pleasant, MI 48859
(989) 774-3576
regan m regan m
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: hair

In reply to this post by Elaine Kunze
Hii
 
Just try to SEM and a simple surgical blade to section hair,
Let me know teh purpose i will be able to suggest you something
I work a lot on hair cross sections and Imaging hair bulbs etc
 
Rgds

On Thu, Dec 4, 2008 at 2:57 AM, Elaine Kunze <[hidden email]> wrote:
Does anyone have suggestions for imaging hair?  We have someone who wants to get a cross section.  Is there some fluorescent label?  I am having difficulty with the density of the hair (light doesn't penetrate very far).

Elaine Kunze
Cytometry Facility
Huck Institute of the Life Sciences
319 Life Sciences Building
Penn State University
University Park, PA 16802
http://www.huck.psu.edu/facilities/cytometry-up/
814-863-2762