![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I would guess that almost 100% of all consumed electric power ultimately goes to heat. Mike -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C Sent: Wednesday, June 29, 2011 10:28 AM To: [hidden email] Subject: heat generation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu |
![]() |
In reply to this post by Watkins, Simon C
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** In my experience the ideal room temperature for the Microscope room is about 72F. Our systems never failed due to heat generation in the room. Also, I have to mention that the argon-ion laser installed in 2001 still working in the Biorad system without any trouble. Of course the laser power went down, but still works. I believe that the temperature maintenance in the microscope room helps a lot in the operation of the high-end microscope systems. Dr. Ammasi Periasamy Professor & Center Director Keck Center for Cellular Imaging (KCCI) Biology, Gilmer Hall (064), 485 McCormick Rd University of Virginia Charlottesville, VA 22904 (Campus Mail - P.O. Box 400328) Voice: 434-243-7602 (Office); 982-4869 (lab) Fax:434-982-5210; Email:[hidden email] http://www.kcci.virginia.edu/Contact/peri.php ************************ 11th Annual Workshop on FRET Microscopy, March 6-10, 2012 http://www.kcci.virginia.edu/workshop/workshop2012/ ************************* -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C Sent: Wednesday, June 29, 2011 10:28 AM To: [hidden email] Subject: heat generation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** To clarify, I am hoping to find out how much heat is put out by the microscopes, not power used (which is an approximation) or what temp the room should be at. As we all know gas lasers pump out thousands of BTU (between 4k and 6k/hour) mercury sources pump out between 1k and 2k BTU. These are the reasons we need to spend so much on air handling for scope rooms. Equally all the old computers/monitors/hardware all produced buckets of heat... Nowadays with all diode lasers, all diode light sources, LCD displays, efficient scan heads etc the amount of heat produced by microscopes has gone down enormously. The numbers from vendors are very conservative, and are probably not actually measured.... so does anyone have any real numbers? S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Periasamy, Ammasi (ap3t) Sent: Wednesday, June 29, 2011 10:49 AM To: [hidden email] Subject: Re: heat generation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** In my experience the ideal room temperature for the Microscope room is about 72F. Our systems never failed due to heat generation in the room. Also, I have to mention that the argon-ion laser installed in 2001 still working in the Biorad system without any trouble. Of course the laser power went down, but still works. I believe that the temperature maintenance in the microscope room helps a lot in the operation of the high-end microscope systems. Dr. Ammasi Periasamy Professor & Center Director Keck Center for Cellular Imaging (KCCI) Biology, Gilmer Hall (064), 485 McCormick Rd University of Virginia Charlottesville, VA 22904 (Campus Mail - P.O. Box 400328) Voice: 434-243-7602 (Office); 982-4869 (lab) Fax:434-982-5210; Email:[hidden email] http://www.kcci.virginia.edu/Contact/peri.php ************************ 11th Annual Workshop on FRET Microscopy, March 6-10, 2012 http://www.kcci.virginia.edu/workshop/workshop2012/ ************************* -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C Sent: Wednesday, June 29, 2011 10:28 AM To: [hidden email] Subject: heat generation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu |
![]() ![]() |
Boswell, Carl A - (cboswell) |
![]() |
In reply to this post by Watkins, Simon C
Hi Simon,
My experience with setting up new systems is that wattage has a direct relationship to heat production, even though the labels may be for peak power consumption. Turnkey systems will provide you with a total wattage of the system, or you can go through each item and total it up. Either way, total watts gives the heating/cooling guys a reasonable figure to work with. Local heat sources, such as argon lasers or multiple power sources/computers in a rack can be selectively vented to avoid their contribution to the heat load in the entire room. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology Univ. of Arizona 520-954-7053 FAX 520-621-3709 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C Sent: Wednesday, June 29, 2011 7:28 AM To: [hidden email] Subject: heat generation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). S Simon C. Watkins Ph.D Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 412-352-2277 www.cbi.pitt.edu |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I've found that the stated current requirements are usually much higher than run time use, and even startup draw. Our old Bio-Rad's electronics added up to about 16 A on a 20A circuit and ocassionally tripped the breaker. Our electricians put an ammeter on the breaker and found that even with powering up that monster CRT display, it never drew more than 4 A. the problem was a surge suppressor with strange harmonics. The specs for the lasers on our FV-1000 suggested that separate 2000 VAC line conditioners were needed for the Ar ion laser and the rest of the lasers. Actual loads were 54% and 5%, respectively, according the conditioner readouts. Has anyone tried the home power usage monitors like the Kill-A-Watt to monitor actual power usage? Glen Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Box 357923 University of Washington Seattle, WA 98195-7923 (206) 616-4156 On Wed, 29 Jun 2011, Boswell, Carl A - (cboswell) wrote: > Hi Simon, > > My experience with setting up new systems is that wattage has a direct relationship to heat production, even though the labels may be for peak power consumption. Turnkey systems will provide you with a total wattage of the system, or you can go through each item and total it up. Either way, total watts gives the heating/cooling guys a reasonable figure to work with. Local heat sources, such as argon lasers or multiple power sources/computers in a rack can be selectively vented to avoid their contribution to the heat load in the entire room. > C > > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > Univ. of Arizona > 520-954-7053 > FAX 520-621-3709 > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Watkins, Simon C > Sent: Wednesday, June 29, 2011 7:28 AM > To: [hidden email] > Subject: heat generation > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all. I am putting together room specs for new microscopes. Has anyone measured real heat production from all diode based confocals (no argon lasers, no mercury bulbs). Actually real heat loads from any microscopes would be interesting (probably to many of us). > S > > Simon C. Watkins Ph.D > Professor and Vice Chair Dept of Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Pittsburgh PA 15261 > 412-352-2277 > www.cbi.pitt.edu > > |
![]() ![]() |
Barbara Foster |
![]() |
In reply to this post by Watkins, Simon C
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** This has been an interesting thread. To take the question one step further... have any of you measured the impact of thermal drift throughout your systems? I'm interested especially in the impact on the stand itself, the objective, the stage, the camera mount, etc. Thanks, Barbara Foster, President and Sr. Consultant Microscopy/Microscopy Education 7101 Royal Glen Trail, Suite A McKinney TX 75070 P: (972)924-5310 Skype: fostermme W: www.MicroscopyEducation.com At 10:10 AM 6/29/2011, you wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >To clarify, I am hoping to find out how much heat is put out by the >microscopes, not power used (which is an approximation) or what temp >the room should be at. As we all know gas lasers pump out thousands >of BTU (between 4k and 6k/hour) mercury sources pump out between 1k >and 2k BTU. These are the reasons we need to spend so much on air >handling for scope rooms. Equally all the old >computers/monitors/hardware all produced buckets of heat... Nowadays >with all diode lasers, all diode light sources, LCD displays, >efficient scan heads etc the amount of heat produced by microscopes >has gone down enormously. The numbers from vendors are very >conservative, and are probably not actually measured.... so does >anyone have any real numbers? >S > >Simon C. Watkins Ph.D >Professor and Vice Chair Dept of Cell Biology >Professor Immunology >Director Center for Biologic Imaging >University of Pittsburgh >Pittsburgh PA 15261 >412-352-2277 >www.cbi.pitt.edu > > >-----Original Message----- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Periasamy, Ammasi (ap3t) >Sent: Wednesday, June 29, 2011 10:49 AM >To: [hidden email] >Subject: Re: heat generation > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >In my experience the ideal room temperature for the Microscope room >is about 72F. Our systems never failed due to heat generation in the >room. Also, I have to mention that the argon-ion laser installed in >2001 still working in the Biorad system without any trouble. Of >course the laser power went down, but still works. I believe that >the temperature maintenance in the microscope room helps a lot in >the operation of the high-end microscope systems. > >Dr. Ammasi Periasamy >Professor & Center Director >Keck Center for Cellular Imaging (KCCI) >Biology, Gilmer Hall (064), 485 McCormick Rd >University of Virginia >Charlottesville, VA 22904 >(Campus Mail - P.O. Box 400328) >Voice: 434-243-7602 (Office); 982-4869 (lab) >Fax:434-982-5210; Email:[hidden email] >http://www.kcci.virginia.edu/Contact/peri.php >************************ >11th Annual Workshop on FRET Microscopy, March 6-10, 2012 >http://www.kcci.virginia.edu/workshop/workshop2012/ >************************* > >-----Original Message----- >From: Confocal Microscopy List >[mailto:[hidden email]] On Behalf Of Watkins, Simon C >Sent: Wednesday, June 29, 2011 10:28 AM >To: [hidden email] >Subject: heat generation > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi all. I am putting together room specs for new microscopes. Has >anyone measured real heat production from all diode based confocals >(no argon lasers, no mercury bulbs). Actually real heat loads from >any microscopes would be interesting (probably to many of us). >S > >Simon C. Watkins Ph.D >Professor and Vice Chair Dept of Cell Biology >Professor Immunology >Director Center for Biologic Imaging >University of Pittsburgh >Pittsburgh PA 15261 >412-352-2277 >www.cbi.pitt.edu |
Free forum by Nabble | Edit this page |