image multiwell plate on confocal microscope

classic Classic list List threaded Threaded
♦
♦
14 messages Options
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello everyone,

We have a Leica SP5 X confocal microscope. We are planning to image  
96-well plates on it. But so far I haven't found anything to fit the  
plate on the stage.

I'm wondering anyone on the list has experiences with this kind of  
experiment. Any suggestion is welcome.

Thank you!

Yue
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear Yue,
Have you contacted the stage manufacturer?   Most motorized stages can accept plates but you may need to buy an adapter from the manufacturer.  Be aware that the distance between the bottom of the plate and the plate's plastic skirt is a few millimeters and varies between plate brands, and styles within a brand. The skirt requires that you bring the object lens so far up to reach the bottom of the plate that the stage can strike either the imaging lens or an adjacent lens, especially when someone else gets on the system with a slide or dish.  If the stage maker does not offer a plate holder that will lower the plate's position to reduce the problem, but you can usually buy a solid blank insert which can be taken to your local machine shop.  

regards,
Glen
Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]








On Oct 12, 2011, at 8:40 AM, [hidden email] wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello everyone,
>
> We have a Leica SP5 X confocal microscope. We are planning to image 96-well plates on it. But so far I haven't found anything to fit the plate on the stage.
>
> I'm wondering anyone on the list has experiences with this kind of experiment. Any suggestion is welcome.
>
> Thank you!
>
> Yue
Emmanuel Gustin Emmanuel Gustin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by yue_li
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Yue,

No personal experience with it, but Leica and EMBL have jointly
developed a hardware & software package to do that. I've seen a webinar
on it, early last year. Maybe that is still archived somewhere. I
suggest asking Leica...

Best Regards,

Emmanuel


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of [hidden email]
Sent: woensdag 12 oktober 2011 17:40
To: [hidden email]
Subject: image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello everyone,

We have a Leica SP5 X confocal microscope. We are planning to image  
96-well plates on it. But so far I haven't found anything to fit the  
plate on the stage.

I'm wondering anyone on the list has experiences with this kind of  
experiment. Any suggestion is welcome.

Thank you!

Yue
Anda Cornea Anda Cornea
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

There is a special stage insert for 96 well plates, I think you need  
to purchase separately if you don't have already.



On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" <[hidden email]
 > wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Yue,
>
> No personal experience with it, but Leica and EMBL have jointly
> developed a hardware & software package to do that. I've seen a  
> webinar
> on it, early last year. Maybe that is still archived somewhere. I
> suggest asking Leica...
>
> Best Regards,
>
> Emmanuel
>
>
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]]
> On Behalf Of [hidden email]
> Sent: woensdag 12 oktober 2011 17:40
> To: [hidden email]
> Subject: image multiwell plate on confocal microscope
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello everyone,
>
> We have a Leica SP5 X confocal microscope. We are planning to image
> 96-well plates on it. But so far I haven't found anything to fit the
> plate on the stage.
>
> I'm wondering anyone on the list has experiences with this kind of
> experiment. Any suggestion is welcome.
>
> Thank you!
>
> Yue
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by Emmanuel Gustin
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Emmanuel,

I'll see if I can find that information online somewhere.

Thank you.

Yue


Quoting "Gustin, Emmanuel [JRDBE]" <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Yue,
>
> No personal experience with it, but Leica and EMBL have jointly
> developed a hardware & software package to do that. I've seen a webinar
> on it, early last year. Maybe that is still archived somewhere. I
> suggest asking Leica...
>
> Best Regards,
>
> Emmanuel
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of [hidden email]
> Sent: woensdag 12 oktober 2011 17:40
> To: [hidden email]
> Subject: image multiwell plate on confocal microscope
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hello everyone,
>
> We have a Leica SP5 X confocal microscope. We are planning to image
> 96-well plates on it. But so far I haven't found anything to fit the
> plate on the stage.
>
> I'm wondering anyone on the list has experiences with this kind of
> experiment. Any suggestion is welcome.
>
> Thank you!
>
> Yue
>
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by Anda Cornea
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Anda, do you know what that special stage is and where I can purchase it?


Quoting Anda Cornea <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> There is a special stage insert for 96 well plates, I think you need
> to purchase separately if you don't have already.
>
>
>
> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" <[hidden email]
>  > wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi Yue,
>>
>> No personal experience with it, but Leica and EMBL have jointly
>> developed a hardware & software package to do that. I've seen a
>> webinar
>> on it, early last year. Maybe that is still archived somewhere. I
>> suggest asking Leica...
>>
>> Best Regards,
>>
>> Emmanuel
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]]
>> On Behalf Of [hidden email]
>> Sent: woensdag 12 oktober 2011 17:40
>> To: [hidden email]
>> Subject: image multiwell plate on confocal microscope
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello everyone,
>>
>> We have a Leica SP5 X confocal microscope. We are planning to image
>> 96-well plates on it. But so far I haven't found anything to fit the
>> plate on the stage.
>>
>> I'm wondering anyone on the list has experiences with this kind of
>> experiment. Any suggestion is welcome.
>>
>> Thank you!
>>
>> Yue
>
Anda Cornea Anda Cornea
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Pretty sure it's from Leica.  Not in my lab today but I can look up details tomorrow.

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Wednesday, October 12, 2011 9:28 AM
To: [hidden email]
Subject: Re: image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Anda, do you know what that special stage is and where I can purchase it?


Quoting Anda Cornea <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> There is a special stage insert for 96 well plates, I think you need
> to purchase separately if you don't have already.
>
>
>
> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]"
> <[hidden email]  > wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi Yue,
>>
>> No personal experience with it, but Leica and EMBL have jointly
>> developed a hardware & software package to do that. I've seen a
>> webinar on it, early last year. Maybe that is still archived
>> somewhere. I suggest asking Leica...
>>
>> Best Regards,
>>
>> Emmanuel
>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List
>> [mailto:[hidden email]]
>> On Behalf Of [hidden email]
>> Sent: woensdag 12 oktober 2011 17:40
>> To: [hidden email]
>> Subject: image multiwell plate on confocal microscope
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello everyone,
>>
>> We have a Leica SP5 X confocal microscope. We are planning to image
>> 96-well plates on it. But so far I haven't found anything to fit the
>> plate on the stage.
>>
>> I'm wondering anyone on the list has experiences with this kind of
>> experiment. Any suggestion is welcome.
>>
>> Thank you!
>>
>> Yue
>
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by Glen MacDonald-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Glen,

I have contacted Leica, and the rep said he needed to ask around.

Thank you for the warning. I'll definitely be very careful when I  
switch stages.

best,
Yue


Quoting Glen MacDonald <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear Yue,
> Have you contacted the stage manufacturer?   Most motorized stages  
> can accept plates but you may need to buy an adapter from the  
> manufacturer.  Be aware that the distance between the bottom of the  
> plate and the plate's plastic skirt is a few millimeters and varies  
> between plate brands, and styles within a brand. The skirt requires  
> that you bring the object lens so far up to reach the bottom of the  
> plate that the stage can strike either the imaging lens or an  
> adjacent lens, especially when someone else gets on the system with  
> a slide or dish.  If the stage maker does not offer a plate holder  
> that will lower the plate's position to reduce the problem, but you  
> can usually buy a solid blank insert which can be taken to your  
> local machine shop.
>
> regards,
> Glen
> Glen MacDonald
> Core for Communication Research
> Virginia Merrill Bloedel Hearing Research Center
> Box 357923
> University of Washington
> Seattle, WA 98195-7923  USA
> (206) 616-4156
> [hidden email]
>
>
>
>
>
>
>
>
> On Oct 12, 2011, at 8:40 AM, [hidden email] wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello everyone,
>>
>> We have a Leica SP5 X confocal microscope. We are planning to image  
>>  96-well plates on it. But so far I haven't found anything to fit  
>> the plate on the stage.
>>
>> I'm wondering anyone on the list has experiences with this kind of  
>> experiment. Any suggestion is welcome.
>>
>> Thank you!
>>
>> Yue
>
Vladimir Ghukasyan-2 Vladimir Ghukasyan-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by Anda Cornea
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Yue,

You need to check the model of your stage then only it will be
possible to search for the 96-well plate holder insert.

Best wishes,
Vladimir


On Wed, Oct 12, 2011 at 12:35 PM, Anda Cornea <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Pretty sure it's from Leica.  Not in my lab today but I can look up details tomorrow.
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
> Sent: Wednesday, October 12, 2011 9:28 AM
> To: [hidden email]
> Subject: Re: image multiwell plate on confocal microscope
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Anda, do you know what that special stage is and where I can purchase it?
>
>
> Quoting Anda Cornea <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> There is a special stage insert for 96 well plates, I think you need
>> to purchase separately if you don't have already.
>>
>>
>>
>> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]"
>> <[hidden email]  > wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Yue,
>>>
>>> No personal experience with it, but Leica and EMBL have jointly
>>> developed a hardware & software package to do that. I've seen a
>>> webinar on it, early last year. Maybe that is still archived
>>> somewhere. I suggest asking Leica...
>>>
>>> Best Regards,
>>>
>>> Emmanuel
>>>
>>>
>>> -----Original Message-----
>>> From: Confocal Microscopy List
>>> [mailto:[hidden email]]
>>> On Behalf Of [hidden email]
>>> Sent: woensdag 12 oktober 2011 17:40
>>> To: [hidden email]
>>> Subject: image multiwell plate on confocal microscope
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hello everyone,
>>>
>>> We have a Leica SP5 X confocal microscope. We are planning to image
>>> 96-well plates on it. But so far I haven't found anything to fit the
>>> plate on the stage.
>>>
>>> I'm wondering anyone on the list has experiences with this kind of
>>> experiment. Any suggestion is welcome.
>>>
>>> Thank you!
>>>
>>> Yue
>>
>
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by Anda Cornea
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Would you please find out the details, Anda? I'd appreciate it very much!

best,
Yue

Quoting Anda Cornea <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Pretty sure it's from Leica.  Not in my lab today but I can look up  
> details tomorrow.
>
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On Behalf Of  
> [hidden email]
> Sent: Wednesday, October 12, 2011 9:28 AM
> To: [hidden email]
> Subject: Re: image multiwell plate on confocal microscope
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Anda, do you know what that special stage is and where I can purchase it?
>
>
> Quoting Anda Cornea <[hidden email]>:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> There is a special stage insert for 96 well plates, I think you need
>> to purchase separately if you don't have already.
>>
>>
>>
>> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]"
>> <[hidden email]  > wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hi Yue,
>>>
>>> No personal experience with it, but Leica and EMBL have jointly
>>> developed a hardware & software package to do that. I've seen a
>>> webinar on it, early last year. Maybe that is still archived
>>> somewhere. I suggest asking Leica...
>>>
>>> Best Regards,
>>>
>>> Emmanuel
>>>
>>>
>>> -----Original Message-----
>>> From: Confocal Microscopy List
>>> [mailto:[hidden email]]
>>> On Behalf Of [hidden email]
>>> Sent: woensdag 12 oktober 2011 17:40
>>> To: [hidden email]
>>> Subject: image multiwell plate on confocal microscope
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hello everyone,
>>>
>>> We have a Leica SP5 X confocal microscope. We are planning to image
>>> 96-well plates on it. But so far I haven't found anything to fit the
>>> plate on the stage.
>>>
>>> I'm wondering anyone on the list has experiences with this kind of
>>> experiment. Any suggestion is welcome.
>>>
>>> Thank you!
>>>
>>> Yue
>>
>
Emmanuel Gustin Emmanuel Gustin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by yue_li
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


Hello,

- Use plates specifically designed for high-content imaging. Typically these have a low skirt,
to avoid a conflict with the objective at the sides of the plate, and a decent optical quality
bottom of the correct thickness.

- On our Leica SP2, we can insert a plate holder in the (Marzhauser) stage but the scanning
range of the stage is not large enough to be able to access the entire plate. I am not sure
whether that is a physical limit or a software limit. It is worth checking this before you
purchase a plate adapter. (We did not, but we almost never use the plate adapter anyway.)

- As far as I know, plate adapters are not compatible with Leica's on-stage Z-scan unit.

Best Regards,

Emmanuel


-----Original Message-----
From: Confocal Microscopy List on behalf of [hidden email]
Sent: Wed 2011-10-12 6:36 PM
To: [hidden email]
Subject: Re: image multiwell plate on confocal microscope
 
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Glen,

I have contacted Leica, and the rep said he needed to ask around.

Thank you for the warning. I'll definitely be very careful when I  
switch stages.

best,
Yue


Quoting Glen MacDonald <[hidden email]>:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear Yue,
> Have you contacted the stage manufacturer?   Most motorized stages  
> can accept plates but you may need to buy an adapter from the  
> manufacturer.  Be aware that the distance between the bottom of the  
> plate and the plate's plastic skirt is a few millimeters and varies  
> between plate brands, and styles within a brand. The skirt requires  
> that you bring the object lens so far up to reach the bottom of the  
> plate that the stage can strike either the imaging lens or an  
> adjacent lens, especially when someone else gets on the system with  
> a slide or dish.  If the stage maker does not offer a plate holder  
> that will lower the plate's position to reduce the problem, but you  
> can usually buy a solid blank insert which can be taken to your  
> local machine shop.
>
> regards,
> Glen
> Glen MacDonald
> Core for Communication Research
> Virginia Merrill Bloedel Hearing Research Center
> Box 357923
> University of Washington
> Seattle, WA 98195-7923  USA
> (206) 616-4156
> [hidden email]
>
>
>
>
>
>
>
>
> On Oct 12, 2011, at 8:40 AM, [hidden email] wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hello everyone,
>>
>> We have a Leica SP5 X confocal microscope. We are planning to image  
>>  96-well plates on it. But so far I haven't found anything to fit  
>> the plate on the stage.
>>
>> I'm wondering anyone on the list has experiences with this kind of  
>> experiment. Any suggestion is welcome.
>>
>> Thank you!
>>
>> Yue
>
Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by yue_li
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I have a Leica SP5X with a Marzhauser motorized stage and a piezo Z-stage.
We bought a separate multi-well plate holder from Leica after the initial
purchase.
This works fine with the piezo z-stage. I haven't directly compared to confirm,
but I am fairly certain the z-range is similar to other stage inserts (just
depends on your sample).
If your stage travel range does not allow you to visit the whole area of the
multi-well plate, this can be adjusted in the software control of the microscope
and stage. Ask your Leica rep or contact me off-line.
I would also be happy to let you know the part number that I ordered from Leica.

- Lisa



---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy
Dana Farber Cancer Institute
Boston, MA
[hidden email]

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of [hidden email]
Sent: Wednesday, October 12, 2011 11:40 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] image multiwell plate on confocal microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello everyone,

We have a Leica SP5 X confocal microscope. We are planning to image  
96-well plates on it. But so far I haven't found anything to fit the  
plate on the stage.

I'm wondering anyone on the list has experiences with this kind of  
experiment. Any suggestion is welcome.

Thank you!

Yue


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
Pascal Weber Pascal Weber
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by yue_li
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I'm really sure you will find the two insert system for your mmultiwell
plate.Directly from Leica. I bought them are:


Insert SuperZ     No : 11640416
For the Marzhauser plate  No : 11531434

Regards
Jeremy Jeremy
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image multiwell plate on confocal microscope

In reply to this post by yue_li
Hi Yue,

There is the TISSUEscope slide scanner - that has confocal fluorescence imaging.  It has a very large stage because it can image large specimens/well plates.  It can hold a glass slide as large as 6x8".  www.huron-technologies.com

JT