				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello everyone, We have a Leica SP5 X confocal microscope. We are planning to image 96-well plates on it. But so far I haven't found anything to fit the plate on the stage. I'm wondering anyone on the list has experiences with this kind of experiment. Any suggestion is welcome. Thank you! Yue  | 
			
 
	
	 | 
						Glen MacDonald-2 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Yue, Have you contacted the stage manufacturer? Most motorized stages can accept plates but you may need to buy an adapter from the manufacturer. Be aware that the distance between the bottom of the plate and the plate's plastic skirt is a few millimeters and varies between plate brands, and styles within a brand. The skirt requires that you bring the object lens so far up to reach the bottom of the plate that the stage can strike either the imaging lens or an adjacent lens, especially when someone else gets on the system with a slide or dish. If the stage maker does not offer a plate holder that will lower the plate's position to reduce the problem, but you can usually buy a solid blank insert which can be taken to your local machine shop. regards, Glen Glen MacDonald Core for Communication Research Virginia Merrill Bloedel Hearing Research Center Box 357923 University of Washington Seattle, WA 98195-7923 USA (206) 616-4156 [hidden email] On Oct 12, 2011, at 8:40 AM, [hidden email] wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everyone, > > We have a Leica SP5 X confocal microscope. We are planning to image 96-well plates on it. But so far I haven't found anything to fit the plate on the stage. > > I'm wondering anyone on the list has experiences with this kind of experiment. Any suggestion is welcome. > > Thank you! > > Yue  | 
			
 
	
	 | 
						Emmanuel Gustin | 
| 
					
		 | 
				
					
	 
				In reply to this post by yue_li
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Yue, No personal experience with it, but Leica and EMBL have jointly developed a hardware & software package to do that. I've seen a webinar on it, early last year. Maybe that is still archived somewhere. I suggest asking Leica... Best Regards, Emmanuel -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: woensdag 12 oktober 2011 17:40 To: [hidden email] Subject: image multiwell plate on confocal microscope ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello everyone, We have a Leica SP5 X confocal microscope. We are planning to image 96-well plates on it. But so far I haven't found anything to fit the plate on the stage. I'm wondering anyone on the list has experiences with this kind of experiment. Any suggestion is welcome. Thank you! Yue  | 
			
 
	
	 | 
						Anda Cornea | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** There is a special stage insert for 96 well plates, I think you need to purchase separately if you don't have already. On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" <[hidden email] > wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Yue, > > No personal experience with it, but Leica and EMBL have jointly > developed a hardware & software package to do that. I've seen a > webinar > on it, early last year. Maybe that is still archived somewhere. I > suggest asking Leica... > > Best Regards, > > Emmanuel > > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of [hidden email] > Sent: woensdag 12 oktober 2011 17:40 > To: [hidden email] > Subject: image multiwell plate on confocal microscope > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everyone, > > We have a Leica SP5 X confocal microscope. We are planning to image > 96-well plates on it. But so far I haven't found anything to fit the > plate on the stage. > > I'm wondering anyone on the list has experiences with this kind of > experiment. Any suggestion is welcome. > > Thank you! > > Yue  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Emmanuel Gustin
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Emmanuel, I'll see if I can find that information online somewhere. Thank you. Yue Quoting "Gustin, Emmanuel [JRDBE]" <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Yue, > > No personal experience with it, but Leica and EMBL have jointly > developed a hardware & software package to do that. I've seen a webinar > on it, early last year. Maybe that is still archived somewhere. I > suggest asking Leica... > > Best Regards, > > Emmanuel > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of [hidden email] > Sent: woensdag 12 oktober 2011 17:40 > To: [hidden email] > Subject: image multiwell plate on confocal microscope > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello everyone, > > We have a Leica SP5 X confocal microscope. We are planning to image > 96-well plates on it. But so far I haven't found anything to fit the > plate on the stage. > > I'm wondering anyone on the list has experiences with this kind of > experiment. Any suggestion is welcome. > > Thank you! > > Yue >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Anda Cornea
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Anda, do you know what that special stage is and where I can purchase it? Quoting Anda Cornea <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > There is a special stage insert for 96 well plates, I think you need > to purchase separately if you don't have already. > > > > On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" <[hidden email] > > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi Yue, >> >> No personal experience with it, but Leica and EMBL have jointly >> developed a hardware & software package to do that. I've seen a >> webinar >> on it, early last year. Maybe that is still archived somewhere. I >> suggest asking Leica... >> >> Best Regards, >> >> Emmanuel >> >> >> -----Original Message----- >> From: Confocal Microscopy List >> [mailto:[hidden email]] >> On Behalf Of [hidden email] >> Sent: woensdag 12 oktober 2011 17:40 >> To: [hidden email] >> Subject: image multiwell plate on confocal microscope >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello everyone, >> >> We have a Leica SP5 X confocal microscope. We are planning to image >> 96-well plates on it. But so far I haven't found anything to fit the >> plate on the stage. >> >> I'm wondering anyone on the list has experiences with this kind of >> experiment. Any suggestion is welcome. >> >> Thank you! >> >> Yue >  | 
			
 
	
	 | 
						Anda Cornea | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Pretty sure it's from Leica. Not in my lab today but I can look up details tomorrow. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Wednesday, October 12, 2011 9:28 AM To: [hidden email] Subject: Re: image multiwell plate on confocal microscope ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Anda, do you know what that special stage is and where I can purchase it? Quoting Anda Cornea <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > There is a special stage insert for 96 well plates, I think you need > to purchase separately if you don't have already. > > > > On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" > <[hidden email] > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi Yue, >> >> No personal experience with it, but Leica and EMBL have jointly >> developed a hardware & software package to do that. I've seen a >> webinar on it, early last year. Maybe that is still archived >> somewhere. I suggest asking Leica... >> >> Best Regards, >> >> Emmanuel >> >> >> -----Original Message----- >> From: Confocal Microscopy List >> [mailto:[hidden email]] >> On Behalf Of [hidden email] >> Sent: woensdag 12 oktober 2011 17:40 >> To: [hidden email] >> Subject: image multiwell plate on confocal microscope >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello everyone, >> >> We have a Leica SP5 X confocal microscope. We are planning to image >> 96-well plates on it. But so far I haven't found anything to fit the >> plate on the stage. >> >> I'm wondering anyone on the list has experiences with this kind of >> experiment. Any suggestion is welcome. >> >> Thank you! >> >> Yue >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Glen MacDonald-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Glen, I have contacted Leica, and the rep said he needed to ask around. Thank you for the warning. I'll definitely be very careful when I switch stages. best, Yue Quoting Glen MacDonald <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Yue, > Have you contacted the stage manufacturer? Most motorized stages > can accept plates but you may need to buy an adapter from the > manufacturer. Be aware that the distance between the bottom of the > plate and the plate's plastic skirt is a few millimeters and varies > between plate brands, and styles within a brand. The skirt requires > that you bring the object lens so far up to reach the bottom of the > plate that the stage can strike either the imaging lens or an > adjacent lens, especially when someone else gets on the system with > a slide or dish. If the stage maker does not offer a plate holder > that will lower the plate's position to reduce the problem, but you > can usually buy a solid blank insert which can be taken to your > local machine shop. > > regards, > Glen > Glen MacDonald > Core for Communication Research > Virginia Merrill Bloedel Hearing Research Center > Box 357923 > University of Washington > Seattle, WA 98195-7923 USA > (206) 616-4156 > [hidden email] > > > > > > > > > On Oct 12, 2011, at 8:40 AM, [hidden email] wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello everyone, >> >> We have a Leica SP5 X confocal microscope. We are planning to image >> 96-well plates on it. But so far I haven't found anything to fit >> the plate on the stage. >> >> I'm wondering anyone on the list has experiences with this kind of >> experiment. Any suggestion is welcome. >> >> Thank you! >> >> Yue >  | 
			
 
	
	 | 
						Vladimir Ghukasyan-2 | 
| 
					
		 | 
				
					
	 
				In reply to this post by Anda Cornea
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Yue, You need to check the model of your stage then only it will be possible to search for the 96-well plate holder insert. Best wishes, Vladimir On Wed, Oct 12, 2011 at 12:35 PM, Anda Cornea <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Pretty sure it's from Leica. Not in my lab today but I can look up details tomorrow. > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] > Sent: Wednesday, October 12, 2011 9:28 AM > To: [hidden email] > Subject: Re: image multiwell plate on confocal microscope > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Anda, do you know what that special stage is and where I can purchase it? > > > Quoting Anda Cornea <[hidden email]>: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> There is a special stage insert for 96 well plates, I think you need >> to purchase separately if you don't have already. >> >> >> >> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" >> <[hidden email] > wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi Yue, >>> >>> No personal experience with it, but Leica and EMBL have jointly >>> developed a hardware & software package to do that. I've seen a >>> webinar on it, early last year. Maybe that is still archived >>> somewhere. I suggest asking Leica... >>> >>> Best Regards, >>> >>> Emmanuel >>> >>> >>> -----Original Message----- >>> From: Confocal Microscopy List >>> [mailto:[hidden email]] >>> On Behalf Of [hidden email] >>> Sent: woensdag 12 oktober 2011 17:40 >>> To: [hidden email] >>> Subject: image multiwell plate on confocal microscope >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hello everyone, >>> >>> We have a Leica SP5 X confocal microscope. We are planning to image >>> 96-well plates on it. But so far I haven't found anything to fit the >>> plate on the stage. >>> >>> I'm wondering anyone on the list has experiences with this kind of >>> experiment. Any suggestion is welcome. >>> >>> Thank you! >>> >>> Yue >> >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Anda Cornea
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Would you please find out the details, Anda? I'd appreciate it very much! best, Yue Quoting Anda Cornea <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Pretty sure it's from Leica. Not in my lab today but I can look up > details tomorrow. > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of > [hidden email] > Sent: Wednesday, October 12, 2011 9:28 AM > To: [hidden email] > Subject: Re: image multiwell plate on confocal microscope > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Anda, do you know what that special stage is and where I can purchase it? > > > Quoting Anda Cornea <[hidden email]>: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> There is a special stage insert for 96 well plates, I think you need >> to purchase separately if you don't have already. >> >> >> >> On Oct 12, 2011, at 9:04 AM, "Gustin, Emmanuel [JRDBE]" >> <[hidden email] > wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi Yue, >>> >>> No personal experience with it, but Leica and EMBL have jointly >>> developed a hardware & software package to do that. I've seen a >>> webinar on it, early last year. Maybe that is still archived >>> somewhere. I suggest asking Leica... >>> >>> Best Regards, >>> >>> Emmanuel >>> >>> >>> -----Original Message----- >>> From: Confocal Microscopy List >>> [mailto:[hidden email]] >>> On Behalf Of [hidden email] >>> Sent: woensdag 12 oktober 2011 17:40 >>> To: [hidden email] >>> Subject: image multiwell plate on confocal microscope >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hello everyone, >>> >>> We have a Leica SP5 X confocal microscope. We are planning to image >>> 96-well plates on it. But so far I haven't found anything to fit the >>> plate on the stage. >>> >>> I'm wondering anyone on the list has experiences with this kind of >>> experiment. Any suggestion is welcome. >>> >>> Thank you! >>> >>> Yue >> >  | 
			
 
	
	 | 
						Emmanuel Gustin | 
| 
					
		 | 
				
					
	 
				In reply to this post by yue_li
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, - Use plates specifically designed for high-content imaging. Typically these have a low skirt, to avoid a conflict with the objective at the sides of the plate, and a decent optical quality bottom of the correct thickness. - On our Leica SP2, we can insert a plate holder in the (Marzhauser) stage but the scanning range of the stage is not large enough to be able to access the entire plate. I am not sure whether that is a physical limit or a software limit. It is worth checking this before you purchase a plate adapter. (We did not, but we almost never use the plate adapter anyway.) - As far as I know, plate adapters are not compatible with Leica's on-stage Z-scan unit. Best Regards, Emmanuel -----Original Message----- From: Confocal Microscopy List on behalf of [hidden email] Sent: Wed 2011-10-12 6:36 PM To: [hidden email] Subject: Re: image multiwell plate on confocal microscope ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Glen, I have contacted Leica, and the rep said he needed to ask around. Thank you for the warning. I'll definitely be very careful when I switch stages. best, Yue Quoting Glen MacDonald <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Yue, > Have you contacted the stage manufacturer? Most motorized stages > can accept plates but you may need to buy an adapter from the > manufacturer. Be aware that the distance between the bottom of the > plate and the plate's plastic skirt is a few millimeters and varies > between plate brands, and styles within a brand. The skirt requires > that you bring the object lens so far up to reach the bottom of the > plate that the stage can strike either the imaging lens or an > adjacent lens, especially when someone else gets on the system with > a slide or dish. If the stage maker does not offer a plate holder > that will lower the plate's position to reduce the problem, but you > can usually buy a solid blank insert which can be taken to your > local machine shop. > > regards, > Glen > Glen MacDonald > Core for Communication Research > Virginia Merrill Bloedel Hearing Research Center > Box 357923 > University of Washington > Seattle, WA 98195-7923 USA > (206) 616-4156 > [hidden email] > > > > > > > > > On Oct 12, 2011, at 8:40 AM, [hidden email] wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello everyone, >> >> We have a Leica SP5 X confocal microscope. We are planning to image >> 96-well plates on it. But so far I haven't found anything to fit >> the plate on the stage. >> >> I'm wondering anyone on the list has experiences with this kind of >> experiment. Any suggestion is welcome. >> >> Thank you! >> >> Yue >  | 
			
 
	
	 | 
						Cameron, Lisa | 
| 
					
		 | 
				
					
	 
				In reply to this post by yue_li
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have a Leica SP5X with a Marzhauser motorized stage and a piezo Z-stage. We bought a separate multi-well plate holder from Leica after the initial purchase. This works fine with the piezo z-stage. I haven't directly compared to confirm, but I am fairly certain the z-range is similar to other stage inserts (just depends on your sample). If your stage travel range does not allow you to visit the whole area of the multi-well plate, this can be adjusted in the software control of the microscope and stage. Ask your Leica rep or contact me off-line. I would also be happy to let you know the part number that I ordered from Leica. - Lisa --------------------------------------- Lisa Cameron, Ph.D. Director of Confocal and Light Microscopy Dana Farber Cancer Institute Boston, MA [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Wednesday, October 12, 2011 11:40 AM To: [hidden email] Subject: [CONFOCALMICROSCOPY] image multiwell plate on confocal microscope ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello everyone, We have a Leica SP5 X confocal microscope. We are planning to image 96-well plates on it. But so far I haven't found anything to fit the plate on the stage. I'm wondering anyone on the list has experiences with this kind of experiment. Any suggestion is welcome. Thank you! Yue The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail.  | 
			
 
	
	 | 
						Pascal Weber | 
| 
					
		 | 
				
					
	 
				In reply to this post by yue_li
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I'm really sure you will find the two insert system for your mmultiwell plate.Directly from Leica. I bought them are: Insert SuperZ No : 11640416 For the Marzhauser plate No : 11531434 Regards  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by yue_li
			 
	
		Hi Yue,
 
				There is the TISSUEscope slide scanner - that has confocal fluorescence imaging. It has a very large stage because it can image large specimens/well plates. It can hold a glass slide as large as 6x8". www.huron-technologies.com JT  | 
			