I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Judy: They use a formula to convert RGB to gray intensities. I am
not sure what this formula is, but it takes into account weighted
contributions from each channel. So if you have pure red (255) and
nothing else in an RGB image and convert it into an 8-bit gray you get
an intensity not 255 but only 85 (which tells you, by the way, the Red
comes with coefficient 0.3).

Michael Model, Ph.D.
Confocal Microscopy Core
Dpt. Biological Sciences
Kent State University
Kent, OH 44242
tel. 330-672-2874

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Judy Trogadis
Sent: Monday, April 07, 2008 11:03 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a
merged RGB image. I would like to analyze just the red channel, so I
split the channels and obtain B & W images of the individual component.
They all look as bright as the original.

However, if I only have a single red image and would like to convert it
to a B & W image (change mode RGB to grey), for 8-bit analysis in
ImageJ, for example, the resulting image is very dull and fainter.
Furthermore, this effect is much more dramatic with a red image than
with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing (bad math)

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Not 0.3 but 1/3 of course


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Judy Trogadis
Sent: Monday, April 07, 2008 11:03 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a
merged RGB image. I would like to analyze just the red channel, so I
split the channels and obtain B & W images of the individual component.
They all look as bright as the original.

However, if I only have a single red image and would like to convert it
to a B & W image (change mode RGB to grey), for 8-bit analysis in
ImageJ, for example, the resulting image is very dull and fainter.
Furthermore, this effect is much more dramatic with a red image than
with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]
Patrick Van Oostveldt Patrick Van Oostveldt
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hye all,


There is a lot if good information about that subject in the microscopy site.

http://www.microscopyu.com/articles/digitalimaging/colorbalance.html

John Russ, is also a man with lot of practical experience in this  
field. He wrote a book on Image processing and several good reviews on  
color photography.

Bye,


Patrick Van Oostveldt

Quoting Judy Trogadis <[hidden email]>:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I have a simple question, related to an annoying problem.
>
> I have an image of a multilabelled fluorescent specimen, displayed  
> as a merged RGB image. I would like to analyze just the red channel,  
>  so I split the channels and obtain B & W images of the individual  
> component. They all look as bright as the original.
>
> However, if I only have a single red image and would like to convert  
>  it to a B & W image (change mode RGB to grey), for 8-bit analysis  
> in  ImageJ, for example, the resulting image is very dull and  
> fainter.  Furthermore, this effect is much more dramatic with a red  
> image than  with a green image.
>
> What is happening?
> Thanks,
> Judy
>
>
> Judy Trogadis
> Bio-Imaging Coordinator
> St. Michael's Hospital, 7Queen
> 30 Bond St.
> Toronto, ON M5B 1W8, Canada
> ph:  416-864-6060  x6337
> pager: 416-685-9219
> fax: 416-864-6043
> [hidden email]
>



--
Dep. Moleculaire Biotechnologie
Coupure links 653
B 9000 GENT

tel 09 264 5969
fax 09 264 6219
Gudrun Ihrke Gudrun Ihrke
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Judy,

Make sure you treat your red image just the same as any merged image:
start with an RGB image (not indexed color) and work with a single
channel rather than converting the image mode from RGB to grey.

Good luck,
Gudrun


On Apr 7, 2008, at 11:02 AM, Judy Trogadis wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I have a simple question, related to an annoying problem.
>
> I have an image of a multilabelled fluorescent specimen, displayed as
> a merged RGB image. I would like to analyze just the red channel, so I
> split the channels and obtain B & W images of the individual
> component. They all look as bright as the original.
>
> However, if I only have a single red image and would like to convert
> it to a B & W image (change mode RGB to grey), for 8-bit analysis in
> ImageJ, for example, the resulting image is very dull and fainter.
> Furthermore, this effect is much more dramatic with a red image than
> with a green image.
>
> What is happening?
> Thanks,
> Judy
>
>
> Judy Trogadis
> Bio-Imaging Coordinator
> St. Michael's Hospital, 7Queen
> 30 Bond St.
> Toronto, ON M5B 1W8, Canada
> ph:  416-864-6060  x6337
> pager: 416-685-9219
> fax: 416-864-6043
> [hidden email]
>
>

____________________________________________
Gudrun Ihrke, Ph.D.
Research Assistant Professor
Department of Pharmacology (C2025)
Uniformed Services University School of Medicine
4301 Jones Bridge Road
Bethesda, MD  20814

phone office:   (301) 295 3225
phone lab: (301) 295 3221
FAX: (301) 295 3220
email: [hidden email]
Paul Herzmark Paul Herzmark
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Things to investigate in your software:

Your microscope should be able to write individual channels as black and white (pure intensity) images, probably as TIFF files. You should not have to convert an RGB image to its component parts. Can you look at a single channel in black and white? Eyeballs evaluate intensities better in black and white than in color.

How does your software handle more than three channel images?

Unless you need to display more than one color on the same image, don't present fluorescent images in monochromatic color.


--
Paul Herzmark
Specialist
[hidden email]

Department of Molecular and Cell Biology
479 Life Science Addition
University of California, Berkeley
Berkeley, CA 94720-3200
(510) 643-9603
(510) 643-9500 fax

On Mon, Apr 7, 2008 at 8:02 AM, Judy Trogadis <[hidden email]> wrote:
I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]



Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Judy, 

With Image J, if you convert an RGB image to monochrome, the gray value you obtain for any given pixel is just a simple average of the red, green, and blue values. 

If your starting image is an RGB image with just red, you should split the channels with image J, and keep the red channel. It should be exactly the same as the original. The green and blue channels should be black (zero value). If you just convert the RGB to 8-bit, you divide the intensity by three (since you average your red value with the values for green and blue, namely zero and zero). 



--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024






On Apr 7, 2008, at 8:02 AM, Judy Trogadis wrote:

Search the CONFOCAL archive at

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening? 
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043

Judy Trogadis Judy Trogadis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Thank you to all who replied - I knew it was simple. Simply converting RGB to 8-bit does not work. It divides the signal and if you have images in all 3 channels, it's total confusion!
I feel embarrassed, I should have known this - I admit it.

Cheers,
Judy

>>> Julio Vazquez <[hidden email]> 04/07/08 1:16 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

-
Judy,

With Image J, if you convert an RGB image to monochrome, the gray  
value you obtain for any given pixel is just a simple average of the  
red, green, and blue values.

If your starting image is an RGB image with just red, you should  
split the channels with image J, and keep the red channel. It should  
be exactly the same as the original. The green and blue channels  
should be black (zero value). If you just convert the RGB to 8-bit,  
you divide the intensity by three (since you average your red value  
with the values for green and blue, namely zero and zero).



--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024






On Apr 7, 2008, at 8:02 AM, Judy Trogadis wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
>
> I have a simple question, related to an annoying problem.
>
> I have an image of a multilabelled fluorescent specimen, displayed  
> as a merged RGB image. I would like to analyze just the red  
> channel, so I split the channels and obtain B & W images of the  
> individual component. They all look as bright as the original.
>
> However, if I only have a single red image and would like to  
> convert it to a B & W image (change mode RGB to grey), for 8-bit  
> analysis in ImageJ, for example, the resulting image is very dull  
> and fainter. Furthermore, this effect is much more dramatic with a  
> red image than with a green image.
>
> What is happening?
> Thanks,
> Judy
>
>
> Judy Trogadis
> Bio-Imaging Coordinator
> St. Michael's Hospital, 7Queen
> 30 Bond St.
> Toronto, ON M5B 1W8, Canada
> ph:  416-864-6060  x6337
> pager: 416-685-9219
> fax: 416-864-6043
> [hidden email]
Mollie Lange Mollie Lange
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Don't feel embarrassed.  Your question and the replies got me thinking  and realizing that I needed to bone up on image processing myself.  I went to Amazon.com and ordered the 5th edition of John Russ's Handbook.
 
So thanks for pushing me to review the basics.
 
Mollie Lange
International Center for Spinal Cord Injury
Kennedy Krieger Institute
Baltimore, Maryland
>>> Judy Trogadis <[hidden email]> 4/7/2008 2:17 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Thank you to all who replied - I knew it was simple. Simply converting RGB to 8-bit does not work. It divides the signal and if you have images in all 3 channels, it's total confusion!
I feel embarrassed, I should have known this - I admit it.

Cheers,
Judy

>>> Julio Vazquez <[hidden email]> 04/07/08 1:16 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

-
Judy,

With Image J, if you convert an RGB image to monochrome, the gray 
value you obtain for any given pixel is just a simple average of the 
red, green, and blue values.

If your starting image is an RGB image with just red, you should 
split the channels with image J, and keep the red channel. It should 
be exactly the same as the original. The green and blue channels 
should be black (zero value). If you just convert the RGB to 8-bit, 
you divide the intensity by three (since you average your red value 
with the values for green and blue, namely zero and zero).



--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024






On Apr 7, 2008, at 8:02 AM, Judy Trogadis wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I have a simple question, related to an annoying problem.
>
> I have an image of a multilabelled fluorescent specimen, displayed 
> as a merged RGB image. I would like to analyze just the red 
> channel, so I split the channels and obtain B & W images of the 
> individual component. They all look as bright as the original.
>
> However, if I only have a single red image and would like to 
> convert it to a B & W image (change mode RGB to grey), for 8-bit 
> analysis in ImageJ, for example, the resulting image is very dull 
> and fainter. Furthermore, this effect is much more dramatic with a 
> red image than with a green image.
>
> What is happening?
> Thanks,
> Judy
>
>
> Judy Trogadis
> Bio-Imaging Coordinator
> St. Michael's Hospital, 7Queen
> 30 Bond St.
> Toronto, ON M5B 1W8, Canada
> ph:  416-864-6060  x6337
> pager: 416-685-9219
> fax: 416-864-6043
> [hidden email]


Disclaimer:
The materials in this e-mail are private and may contain Protected Health Information. Please note that e-mail is not necessarily confidential or secure. Your use of e-mail constitutes your acknowledgment of these confidentiality and security limitations. If you are not the intended recipient, be advised that any unauthorized use, disclosure, copying, distribution, or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this e-mail in error, please immediately notify the sender via telephone or return e-mail.
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal image processing
As others have pointed out, this happens because in a simple conversion
of colour to grey-scale red has a low weighting.  But you shouldn't be
doing that ....
 
If your red image is stored as an rgb data set (24 bit) just separate the
channels exactly as you did with your rgb image.  You well get two blank
images for green and blue and the third will be what you want.
 
If your image is stored as paletted colour (8-bit, 256 colours) you can select
'convert to rgb' and then split the channels as above.  Or you can just replace 
the palette with a linear grey palette. Make sure you select 'maintain indexes' or
equivalent, NOT 'use closest value'.  If your imaging programme doesn't have a
linear grey palette in its list, save the palette from a good-quality grey-scale
image to get one.
 
                                                                                                Guy
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of Judy Trogadis
Sent: Tue 08/04/08 1:02 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]

George Ring George Ring
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

image processing
To point out what is probably obvious, I believe that this procedure will only work if working with a maximum of three labels, red, green and blue.  Only then can you separate the merged image into the RGB component channels.  However, Judy wrote that she has multilabeled specimens - perhaps more than three.  Thus, any additional fluorophore would be labeled with a LUT that would be some combination of Red, Green and Blue.  Separating the Red channel of a such an RGB merged image would also separate out the red component of a second or third fluorophore.  Just my 2 cents worth.
 
George
 
 
 
George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Guy Cox <[hidden email]> 4/7/08 8:22:15 PM >>>
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
As others have pointed out, this happens because in a simple conversion
of colour to grey-scale red has a low weighting.  But you shouldn't be
doing that ....
 
If your red image is stored as an rgb data set (24 bit) just separate the
channels exactly as you did with your rgb image.  You well get two blank
images for green and blue and the third will be what you want.
 
If your image is stored as paletted colour (8-bit, 256 colours) you can select
'convert to rgb' and then split the channels as above.  Or you can just replace 
the palette with a linear grey palette. Make sure you select 'maintain indexes' or
equivalent, NOT 'use closest value'.  If your imaging programme doesn't have a
linear grey palette in its list, save the palette from a good-quality grey-scale
image to get one.
 
                                                                                                Guy
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of Judy Trogadis
Sent: Tue 08/04/08 1:02 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]

vb-2 vb-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: image processing

In reply to this post by Guy Cox
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal image processing
Dear List,
 
Does the recent version of Photoshop allow re-mapping to [0,4095], [0,65535], based on zero, linear (Log) stretch, etc.?
Is there a free plug-in to read 12-bit images in Photoshop? I heard that John Russ's handbook had a supllement CD for extra price with lots of different but useful plug-ins, any feedback would be appreciated? 
 
Cheers,
 
Vitaly
NCI-Frederick,
301-846-6575
 
----- Original Message -----
Sent: Monday, April 07, 2008 7:22 PM
Subject: Re: image processing

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
As others have pointed out, this happens because in a simple conversion
of colour to grey-scale red has a low weighting.  But you shouldn't be
doing that ....
 
If your red image is stored as an rgb data set (24 bit) just separate the
channels exactly as you did with your rgb image.  You well get two blank
images for green and blue and the third will be what you want.
 
If your image is stored as paletted colour (8-bit, 256 colours) you can select
'convert to rgb' and then split the channels as above.  Or you can just replace 
the palette with a linear grey palette. Make sure you select 'maintain indexes' or
equivalent, NOT 'use closest value'.  If your imaging programme doesn't have a
linear grey palette in its list, save the palette from a good-quality grey-scale
image to get one.
 
                                                                                                Guy
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of Judy Trogadis
Sent: Tue 08/04/08 1:02 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]

Thomas Pitta Thomas Pitta
Reply | Threaded
Open this post in threaded view
♦
♦
|

remove

In reply to this post by George Ring
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal image processing

Please remove me from the list.  Thankyou

 

Thomas Pitta Ph.D.

Micro Video Instruments

Digital Imaging Specialist

508 580 0080 ex 241

508 330 8356 cell

[hidden email]


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of George Ring
Sent: Tuesday, April 08, 2008 9:43 AM
To: [hidden email]
Subject: Re: image processing

 

To point out what is probably obvious, I believe that this procedure will only work if working with a maximum of three labels, red, green and blue.  Only then can you separate the merged image into the RGB component channels.  However, Judy wrote that she has multilabeled specimens - perhaps more than three.  Thus, any additional fluorophore would be labeled with a LUT that would be some combination of Red, Green and Blue.  Separating the Red channel of a such an RGB merged image would also separate out the red component of a second or third fluorophore.  Just my 2 cents worth.

 

George

 

 

 

George Ring, Ph.D.
Dept. of Cell and Developmental Biology
SUNY Upstate Medical University
750 E. Adams St.
Syracuse NY  13210
Tel. (315) 464-8595
FAX (315) 464-8535
email: [hidden email]

>>> Guy Cox <[hidden email]> 4/7/08 8:22:15 PM >>>

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

As others have pointed out, this happens because in a simple conversion

of colour to grey-scale red has a low weighting.  But you shouldn't be

doing that ....

 

If your red image is stored as an rgb data set (24 bit) just separate the

channels exactly as you did with your rgb image.  You well get two blank

images for green and blue and the third will be what you want.

 

If your image is stored as paletted colour (8-bit, 256 colours) you can select

'convert to rgb' and then split the channels as above.  Or you can just replace 

the palette with a linear grey palette. Make sure you select 'maintain indexes' or

equivalent, NOT 'use closest value'.  If your imaging programme doesn't have a

linear grey palette in its list, save the palette from a good-quality grey-scale

image to get one.

 

                                                                                                Guy

 

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis

     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net

 


From: Confocal Microscopy List on behalf of Judy Trogadis
Sent: Tue 08/04/08 1:02 AM
To: [hidden email]
Subject: image processing

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I have a simple question, related to an annoying problem.

I have an image of a multilabelled fluorescent specimen, displayed as a merged RGB image. I would like to analyze just the red channel, so I split the channels and obtain B & W images of the individual component. They all look as bright as the original.

However, if I only have a single red image and would like to convert it to a B & W image (change mode RGB to grey), for 8-bit analysis in ImageJ, for example, the resulting image is very dull and fainter. Furthermore, this effect is much more dramatic with a red image than with a green image.

What is happening?
Thanks,
Judy


Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]