immersion fluid options for long timelapses with long working distance water objective

classic Classic list List threaded Threaded
♦
♦
15 messages Options
Ariel, Pablo Ariel, Pablo
Reply | Threaded
Open this post in threaded view
♦
♦
|

immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory
Brinkhous-Bullitt Bldg B04
Department of Pathology and Laboratory Medicine
University of North Carolina at Chapel Hill
http://www.med.unc.edu/microscopy
Tel: 919-966-2413
Tilo Moede Tilo Moede
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://aka.ms/AAb9ysg>
________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=rFgQ0fCjIpE%2FNn0EBPQErmRAEpYW6vQ%2BqH37KgH%2FUT0%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory
Brinkhous-Bullitt Bldg B04
Department of Pathology and Laboratory Medicine
University of North Carolina at Chapel Hill
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=4wo53%2Bh3lqdhhPlnEhu2oibKqKHBI3E2kO%2BBpX4wmfo%3D&amp;reserved=0
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
Gary Laevsky Gary Laevsky
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Ariel, Pablo
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Many years ago we used Suave hair gel. It worked for multiple positions fairly close together.

Best,

Gary



> On Apr 15, 2021, at 6:08 AM, Ariel, Pablo <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all,
>
> I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.
>
> While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.
>
> Thanks,
> Pablo
>
> Pablo Ariel, Ph.D.
> Assistant Professor
> Director of the Microscopy Services Laboratory
> Brinkhous-Bullitt Bldg B04
> Department of Pathology and Laboratory Medicine
> University of North Carolina at Chapel Hill
> http://www.med.unc.edu/microscopy
> Tel: 919-966-2413
Ariel, Pablo Ariel, Pablo
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Tilo Moede
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks for the suggestion Tilo.
How long do the viscotears last before evaporating? We are hoping for multiple position imaging over 48 hours.
Best,
Pablo

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Tilo Moede
Sent: Thursday, April 15, 2021 6:15 AM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://aka.ms/AAb9ysg> ________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=rFgQ0fCjIpE%2FNn0EBPQErmRAEpYW6vQ%2BqH37KgH%2FUT0%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=4wo53%2Bh3lqdhhPlnEhu2oibKqKHBI3E2kO%2BBpX4wmfo%3D&amp;reserved=0
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
Cammer, Michael-2 Cammer, Michael-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

A few years ago people on this listserv suggested jelly for sonograms and Genteal eye gel.  Here is info on the latter http://microscopynotes.com/710/genteal/index.html


Michael Cammer, Sr Research Scientist, DART Microscopy Laboratory

NYU Langone Health, 540 First Avenue, SK2 Microscopy Suite, New York, NY  10016

[hidden email]<mailto:[hidden email]>  http://nyulmc.org/micros  http://microscopynotes.com/

Voice direct only, no text or messages:  1-914-309-3270 and 1-646-501-0567



________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 6:41:53 AM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTiX8EklI$
Post images on https://urldefense.com/v3/__http://www.imgur.com__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTutLxWag$  and include the link in your posting.
*****

Thanks for the suggestion Tilo.
How long do the viscotears last before evaporating? We are hoping for multiple position imaging over 48 hours.
Best,
Pablo

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Tilo Moede
Sent: Thursday, April 15, 2021 6:15 AM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.com/v3/__http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTiX8EklI$
Post images on https://urldefense.com/v3/__http://www.imgur.com__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTutLxWag$  and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://urldefense.com/v3/__https://aka.ms/AAb9ysg__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTGHe3E_0$ > ________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8%3D&amp;reserved=0
Post images on https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=rFgQ0fCjIpE%2FNn0EBPQErmRAEpYW6vQ%2BqH37KgH%2FUT0%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://eur01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7CTilo.Moede%40KI.SE%7C3517a286bb174a78639b08d8fff65b64%7Cbff7eef1cf4b4f32be3da1dda043c05d%7C0%7C0%7C637540780863329546%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=4wo53%2Bh3lqdhhPlnEhu2oibKqKHBI3E2kO%2BBpX4wmfo%3D&amp;reserved=0
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.com/v3/__https://ki.se/medarbetare/integritetsskyddspolicy__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTMJLmYaA$ >.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.com/v3/__https://ki.se/en/staff/data-protection-policy__;!!MXfaZl3l!NV0LInRfEiVhwJn3wg12JwYCTat4uHR7sWvEWVnmG_jiEDgNogdg1LQn_Un-qnVTobazN_U$ >.
Tim Feinstein Tim Feinstein
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Ariel, Pablo
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Why not use the Cargille oils that match the refractive index of water?  You can get separate oils for matching water at room temperature (~27C) or incubation temperature (37C).  I used those for long term imaging without any evaporation issues.  I don't remember the product #s but I can look them up if you like.  

All the best,


T

Timothy Feinstein, Ph.D.


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 6:05 AM
To: [hidden email]
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=hA8vGP0sPHELDD2uLakMIMwyzaRT%2FB%2FIzr8W%2BUZTwjQ%3D&amp;reserved=0
Post images on https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=ioe%2BTAGtoI5M2ly81aUixXGRZSSMIMp%2FAUipc0RESoM%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=wnOHApiZUAd6ZVqaqBQCrP7O8wCev%2BND97lhM2FmhQY%3D&amp;reserved=0
Tel: 919-966-2413
Cammer, Michael-2 Cammer, Michael-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Long working distances require that the fluid maintain its structure, such as with high viscosity or strong surface tension.  The oils run down the side of the lens.
Cheers-
Michael

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Feinstein, Timothy N
Sent: Thursday, April 15, 2021 9:55 AM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Why not use the Cargille oils that match the refractive index of water?  You can get separate oils for matching water at room temperature (~27C) or incubation temperature (37C).  I used those for long term imaging without any evaporation issues.  I don't remember the product #s but I can look them up if you like.

All the best,


T

Timothy Feinstein, Ph.D.


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 6:05 AM
To: [hidden email]
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=hA8vGP0sPHELDD2uLakMIMwyzaRT%2FB%2FIzr8W%2BUZTwjQ%3D&amp;reserved=0
Post images on https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=ioe%2BTAGtoI5M2ly81aUixXGRZSSMIMp%2FAUipc0RESoM%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=wnOHApiZUAd6ZVqaqBQCrP7O8wCev%2BND97lhM2FmhQY%3D&amp;reserved=0
Tel: 919-966-2413

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Periasamy, Ammasi (ap3t)-2 Periasamy, Ammasi (ap3t)-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Ariel, Pablo
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Pablo,
Zeiss sells oil with refractive index 1.33.
This will go with any companies water immersion lenses.
we are using for all the microscopy systems for long time time-lapse imaging.
Hope thus helps.
Best,
Ammasi


Dr. Ammasi Periasamy
Director and founder of WM Keck Center for Cellular Imaging,
Prof. of Biology & Biomed. Eng., University of Virginia,
Physical & Life Sciences Building (PLSB 005)
90 Geldard Dr., Charlottesville, VA 22904, USA.

http://www.kcci.virginia.edu/people/profile/ap3t
Phone: (434) 243-7602 or 982-4869
Fax: (434) 982-5210
E-mail: [hidden email]

20th Annual FLIM and FRET Workshop-March 7-11, 2022.
http://www.kcci.virginia.edu/workshop 


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 6:05 AM
To: [hidden email]
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill http://www.med.unc.edu/microscopy
Tel: 919-966-2413
Cammer, Michael-2 Cammer, Michael-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Tim Feinstein
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Long working distances require that the fluid maintain its structure, such as with high viscosity or strong surface tension.  The oils (at least the ones we've tried such as Immersol W) run down the side of the lens.
Cheers-
Michael

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Feinstein, Timothy N
Sent: Thursday, April 15, 2021 9:55 AM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

[EXTERNAL]

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Why not use the Cargille oils that match the refractive index of water?  You can get separate oils for matching water at room temperature (~27C) or incubation temperature (37C).  I used those for long term imaging without any evaporation issues.  I don't remember the product #s but I can look them up if you like.

All the best,


T

Timothy Feinstein, Ph.D.


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 6:05 AM
To: [hidden email]
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=hA8vGP0sPHELDD2uLakMIMwyzaRT%2FB%2FIzr8W%2BUZTwjQ%3D&amp;reserved=0
Post images on https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=ioe%2BTAGtoI5M2ly81aUixXGRZSSMIMp%2FAUipc0RESoM%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=wnOHApiZUAd6ZVqaqBQCrP7O8wCev%2BND97lhM2FmhQY%3D&amp;reserved=0
Tel: 919-966-2413

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
TSwayne TSwayne
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Tilo Moede
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Pablo,

We have used Genteal Severe Gel (Hypromellose 0.3%) with our LWD lens on an inverted system. It lasts overnight, but will not last 48 hours without replenishment. For longer experiments, we re-apply during a break in imaging, roughly every 12 hours or so. We also made a little wall out of a glove finger to hold the gel in place and I think this reduces evaporation as well (but it limits ability to do very large Z stacks, as the glove material folds under pressure).

Hope this helps,
Theresa



------------------------------------
Theresa Swayne, Ph.D.
Associate Research Scientist
Manager, Confocal and Specialized Microscopy Shared Resource<http://hiccc.columbia.edu/research/sharedresources/confocal>
(she/her/hers)

Mobile: 347-920-6865 (anytime)
[hidden email] <mailto:[hidden email]> (Office hours only: 8-7, Mon-Fri)
Core customers: Please contact BOTH Laura Munteanu (elm2157/347-721-7851) AND me for the fastest response.

Herbert Irving Comprehensive Cancer Center
Columbia University Irving Medical Center
Lasker Biomedical Research Building
3960 Broadway, Room 320
New York, NY 10032





From: Confocal Microscopy List <[hidden email]> on behalf of Tilo Moede <[hidden email]>
Date: Thursday, April 15, 2021 at 6:35 AM
To: [hidden email] <[hidden email]>
Subject: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=ZZHOPNNkyr5Ym7HuL68QB_BZnfAYLM-vEAmpLIxDOzc&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=PYfF1poS60mJEIWYsRKSmq6N8Q9R1ARdNRUtS-Lc3WY&e=  and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_AAb9ysg&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=i0M89wHY3rpsoGsdbotuCX1Pus6LNE9KqVCIPdQsbUk&e= >
________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Flists.umn.edu-252Fcgi-2Dbin-252Fwa-253FA0-253Dconfocalmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=onGHE-h_33cMxFnQ5PQj8Lk-1aQH-xDHcMIeOYO8Lvw&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.imgur.com-252F-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3DrFgQ0fCjIpE-252FNn0EBPQErmRAEpYW6vQ-252BqH37KgH-252FUT0-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=4JAb5EswatQ-lYpnFxzP9qk5OLUH4haCyChomrHYWeM&e=  and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory
Brinkhous-Bullitt Bldg B04
Department of Pathology and Laboratory Medicine
University of North Carolina at Chapel Hill
https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.med.unc.edu-252Fmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D4wo53-252Bh3lqdhhPlnEhu2oibKqKHBI3E2kO-252BBpX4wmfo-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=sPnYkSPIO9WVpX90LDSPI5ylwKf1R1ChYz7qGvkTZ64&e=
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=jrIlbDxhTaT0hy1zcEA35R8Aj6E9B1Ndp-dscg50iPs&e= >.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=wIhUo_BGeQ7zzsbSQ_vl-IewAwcVEbKOQmtlzL-7LC0&e= >.
Ariel, Pablo Ariel, Pablo
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks all for the many suggestions received on and off list.

I will try the eyedrop approach and let everyone know if we come up with some sort of containment that can reduce evaporation further.

Best,
Pablo

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Swayne, Theresa C.
Sent: Thursday, April 15, 2021 10:10 AM
To: [hidden email]
Subject: Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Pablo,

We have used Genteal Severe Gel (Hypromellose 0.3%) with our LWD lens on an inverted system. It lasts overnight, but will not last 48 hours without replenishment. For longer experiments, we re-apply during a break in imaging, roughly every 12 hours or so. We also made a little wall out of a glove finger to hold the gel in place and I think this reduces evaporation as well (but it limits ability to do very large Z stacks, as the glove material folds under pressure).

Hope this helps,
Theresa



------------------------------------
Theresa Swayne, Ph.D.
Associate Research Scientist
Manager, Confocal and Specialized Microscopy Shared Resource<http://hiccc.columbia.edu/research/sharedresources/confocal>
(she/her/hers)

Mobile: 347-920-6865 (anytime)
[hidden email] <mailto:[hidden email]> (Office hours only: 8-7, Mon-Fri) Core customers: Please contact BOTH Laura Munteanu (elm2157/347-721-7851) AND me for the fastest response.

Herbert Irving Comprehensive Cancer Center Columbia University Irving Medical Center Lasker Biomedical Research Building
3960 Broadway, Room 320
New York, NY 10032





From: Confocal Microscopy List <[hidden email]> on behalf of Tilo Moede <[hidden email]>
Date: Thursday, April 15, 2021 at 6:35 AM
To: [hidden email] <[hidden email]>
Subject: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=ZZHOPNNkyr5Ym7HuL68QB_BZnfAYLM-vEAmpLIxDOzc&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=PYfF1poS60mJEIWYsRKSmq6N8Q9R1ARdNRUtS-Lc3WY&e=  and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_AAb9ysg&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=i0M89wHY3rpsoGsdbotuCX1Pus6LNE9KqVCIPdQsbUk&e= > ________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Flists.umn.edu-252Fcgi-2Dbin-252Fwa-253FA0-253Dconfocalmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=onGHE-h_33cMxFnQ5PQj8Lk-1aQH-xDHcMIeOYO8Lvw&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.imgur.com-252F-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3DrFgQ0fCjIpE-252FNn0EBPQErmRAEpYW6vQ-252BqH37KgH-252FUT0-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=4JAb5EswatQ-lYpnFxzP9qk5OLUH4haCyChomrHYWeM&e=  and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.med.unc.edu-252Fmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D4wo53-252Bh3lqdhhPlnEhu2oibKqKHBI3E2kO-252BBpX4wmfo-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=sPnYkSPIO9WVpX90LDSPI5ylwKf1R1ChYz7qGvkTZ64&e=
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=jrIlbDxhTaT0hy1zcEA35R8Aj6E9B1Ndp-dscg50iPs&e= >.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=wIhUo_BGeQ7zzsbSQ_vl-IewAwcVEbKOQmtlzL-7LC0&e= >.
Cromey, Douglas W - (dcromey)-2 Cromey, Douglas W - (dcromey)-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [EXT]Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

http://confocal-microscopy-list.588098.n2.nabble.com/water-refill-device-for-WI-objective-td5566734.html

I remember that Steve Cody worked on this problem (I met Steve at Jim Pawley's course back in 2008). See his response in the thread above.

Doug

^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
Douglas W. Cromey, M.S. - Associate Scientific Investigator
Dept. of Cellular & Molecular Medicine, University of Arizona
Life Sciences North, Room 463
1333 N. Martin Ave, Tucson, AZ  85721 USA

office:  LSN 463        email: [hidden email]
voice:  520-626-2824       fax:  520-626-2097

UA Microscopy Alliance - http://microscopy.arizona.edu/
A collaborative effort to bring information about shared microscopy
facilities to the University of Arizona and the community.

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 7:49 AM
To: [hidden email]
Subject: [EXT]Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

External Email

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks all for the many suggestions received on and off list.

I will try the eyedrop approach and let everyone know if we come up with some sort of containment that can reduce evaporation further.

Best,
Pablo

-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Swayne, Theresa C.
Sent: Thursday, April 15, 2021 10:10 AM
To: [hidden email]
Subject: Re: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Pablo,

We have used Genteal Severe Gel (Hypromellose 0.3%) with our LWD lens on an inverted system. It lasts overnight, but will not last 48 hours without replenishment. For longer experiments, we re-apply during a break in imaging, roughly every 12 hours or so. We also made a little wall out of a glove finger to hold the gel in place and I think this reduces evaporation as well (but it limits ability to do very large Z stacks, as the glove material folds under pressure).

Hope this helps,
Theresa



------------------------------------
Theresa Swayne, Ph.D.
Associate Research Scientist
Manager, Confocal and Specialized Microscopy Shared Resource<http://hiccc.columbia.edu/research/sharedresources/confocal>
(she/her/hers)

Mobile: 347-920-6865 (anytime)
[hidden email] <mailto:[hidden email]> (Office hours only: 8-7, Mon-Fri) Core customers: Please contact BOTH Laura Munteanu (elm2157/347-721-7851) AND me for the fastest response.

Herbert Irving Comprehensive Cancer Center Columbia University Irving Medical Center Lasker Biomedical Research Building
3960 Broadway, Room 320
New York, NY 10032





From: Confocal Microscopy List <[hidden email]> on behalf of Tilo Moede <[hidden email]>
Date: Thursday, April 15, 2021 at 6:35 AM
To: [hidden email] <[hidden email]>
Subject: [EXTERNAL] Re: immersion fluid options for long timelapses with long working distance water objective
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=ZZHOPNNkyr5Ym7HuL68QB_BZnfAYLM-vEAmpLIxDOzc&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=PYfF1poS60mJEIWYsRKSmq6N8Q9R1ARdNRUtS-Lc3WY&e=  and include the link in your posting.
*****

We use viscotears(artificial tear-fluid replacement) as immersion fluid with exactly that lens (and it's 10x and 20x relatives) for in vivo imaging of organoids transplanted to the anterior chamber of the eye with great results.. Something like that should be available from a pharmacy.
Cheerio
Tilo


Hämta Outlook för Android<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_AAb9ysg&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=i0M89wHY3rpsoGsdbotuCX1Pus6LNE9KqVCIPdQsbUk&e= > ________________________________
From: Confocal Microscopy List <[hidden email]> on behalf of Ariel, Pablo <[hidden email]>
Sent: Thursday, April 15, 2021 12:05:25 PM
To: [hidden email] <[hidden email]>
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Flists.umn.edu-252Fcgi-2Dbin-252Fwa-253FA0-253Dconfocalmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D2qHml3iWWWFliCmDgStyuaYKjK7jnpvZhSg6vzZxSs8-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=onGHE-h_33cMxFnQ5PQj8Lk-1aQH-xDHcMIeOYO8Lvw&e=
Post images on https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.imgur.com-252F-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3DrFgQ0fCjIpE-252FNn0EBPQErmRAEpYW6vQ-252BqH37KgH-252FUT0-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=4JAb5EswatQ-lYpnFxzP9qk5OLUH4haCyChomrHYWeM&e=  and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill https://urldefense.proofpoint.com/v2/url?u=https-3A__eur01.safelinks.protection.outlook.com_-3Furl-3Dhttp-253A-252F-252Fwww.med.unc.edu-252Fmicroscopy-26amp-3Bdata-3D04-257C01-257CTilo.Moede-2540KI.SE-257C3517a286bb174a78639b08d8fff65b64-257Cbff7eef1cf4b4f32be3da1dda043c05d-257C0-257C0-257C637540780863329546-257CUnknown-257CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0-253D-257C1000-26amp-3Bsdata-3D4wo53-252Bh3lqdhhPlnEhu2oibKqKHBI3E2kO-252BBpX4wmfo-253D-26amp-3Breserved-3D0&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=sPnYkSPIO9WVpX90LDSPI5ylwKf1R1ChYz7qGvkTZ64&e=
Tel: 919-966-2413



När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=jrIlbDxhTaT0hy1zcEA35R8Aj6E9B1Ndp-dscg50iPs&e= >.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI's processing of personal data here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIFAw&c=G2MiLlal7SXE3PeSnG8W6_JBU6FcdVjSsBSbw6gcR0U&r=JzqbFzGMeaMAlwu5KvYtC8ig-FF8xbvhI6BlFtbXky8&m=4gcujy7AjOTziHxbnfuOrypVxO9f7GuDJ08xTw8y8SY&s=wIhUo_BGeQ7zzsbSQ_vl-IewAwcVEbKOQmtlzL-7LC0&e= >.
Jeremy Adler-4 Jeremy Adler-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Tim Feinstein
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Assuming you are using a petri dish with a glass bottom and adherent cells.
Grow you cells on a second coverslip and mount it in the petri dish face down with a 1.5mm spacer underneath.
This should get the objective close enough to the bottom of the petri dish for the oil to occupy the gap.

Jeremy Adler
BioVis
Uppsala


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Feinstein, Timothy N
Sent: Thursday, April 15, 2021 3:55 PM
To: [hidden email]
Subject: Re: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Why not use the Cargille oils that match the refractive index of water?  You can get separate oils for matching water at room temperature (~27C) or incubation temperature (37C).  I used those for long term imaging without any evaporation issues.  I don't remember the product #s but I can look them up if you like.

All the best,


T

Timothy Feinstein, Ph.D.


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Ariel, Pablo
Sent: Thursday, April 15, 2021 6:05 AM
To: [hidden email]
Subject: immersion fluid options for long timelapses with long working distance water objective

*****
To join, leave or search the confocal microscopy listserv, go to:
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=hA8vGP0sPHELDD2uLakMIMwyzaRT%2FB%2FIzr8W%2BUZTwjQ%3D&amp;reserved=0
Post images on https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=ioe%2BTAGtoI5M2ly81aUixXGRZSSMIMp%2FAUipc0RESoM%3D&amp;reserved=0 and include the link in your posting.
*****

Dear all,

I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.

While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.

Thanks,
Pablo

Pablo Ariel, Ph.D.
Assistant Professor
Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04 Department of Pathology and Laboratory Medicine University of North Carolina at Chapel Hill
https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=wnOHApiZUAd6ZVqaqBQCrP7O8wCev%2BND97lhM2FmhQY%3D&amp;reserved=0
Tel: 919-966-2413








När du har kontakt med oss på Uppsala universitet med e-post så innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/

E-mailing Uppsala University means that we will process your personal data. For more information on how this is performed, please read here: http://www.uu.se/en/about-uu/data-protection-policy
Michael Giacomelli-2 Michael Giacomelli-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: [EXT] immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Ariel, Pablo
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I use UV cure adhesive to make a wall around the coverslip and then
fill up the coverslip with a larger volume of water.  It'll eventually
evaporate, but for <24 hours it isn't too hard to have enough water
that evaporation isn't a problem.

It is not an elegant solution but it has always worked well enough.

Michael

On Thu, Apr 15, 2021 at 6:08 AM Ariel, Pablo <[hidden email]> wrote:

>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFAg&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=b0tDWzqI-CVUzLk7XzsOt3qQqQtXQsPpKR__Jwh_go8&s=mZn0L-XgBz4bsBzZdtkcMvCe8mIRaV2xL8RwpW1SIOE&e=
> Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFAg&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=b0tDWzqI-CVUzLk7XzsOt3qQqQtXQsPpKR__Jwh_go8&s=F-0sPOkToi58KLNeOYOsyYK_7bfx-_y3pF3JLIIPpkE&e=  and include the link in your posting.
> *****
>
> Dear all,
>
> I am demoing a long working distance water objective from Leica (HC FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted configuration, and am running into some challenges.
>
> While the objective is very nice, the long working distance presents a problem for long-duration timelapses with multiple positions. Water works fine, but it eventually evaporates (in hours). Zeiss Immersol W does not evaporate, but cannot hold the column when the objective is over a mm from the coverslip (it drips/slides off). Leica has an automated water dispenser, but it is not available for my current demo. I was wondering if anyone had any ideas for other immersion options that have a surface tension and RI similar to water, but don't evaporate as easily.
>
> Thanks,
> Pablo
>
> Pablo Ariel, Ph.D.
> Assistant Professor
> Director of the Microscopy Services Laboratory
> Brinkhous-Bullitt Bldg B04
> Department of Pathology and Laboratory Medicine
> University of North Carolina at Chapel Hill
> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.med.unc.edu_microscopy&d=DwIFAg&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=b0tDWzqI-CVUzLk7XzsOt3qQqQtXQsPpKR__Jwh_go8&s=EYNR-YcjTsR3JTlVsAg8U4yVx_EiDEhR7y1rLbo-YsA&e=
> Tel: 919-966-2413
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: immersion fluid options for long timelapses with long working distance water objective

In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Leica also sells a watering system. One of the cores around here has it on
their SP8. It seems to work fairly well; slowly replenishing water as it
evaporates and wicking away any excess with a foam collar fitting.

Craig

On Thu, Apr 15, 2021 at 10:58 AM Jeremy Adler <[hidden email]>
wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Assuming you are using a petri dish with a glass bottom and adherent cells.
> Grow you cells on a second coverslip and mount it in the petri dish face
> down with a 1.5mm spacer underneath.
> This should get the objective close enough to the bottom of the petri dish
> for the oil to occupy the gap.
>
> Jeremy Adler
> BioVis
> Uppsala
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Feinstein, Timothy N
> Sent: Thursday, April 15, 2021 3:55 PM
> To: [hidden email]
> Subject: Re: immersion fluid options for long timelapses with long working
> distance water objective
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Why not use the Cargille oils that match the refractive index of water?
> You can get separate oils for matching water at room temperature (~27C) or
> incubation temperature (37C).  I used those for long term imaging without
> any evaporation issues.  I don't remember the product #s but I can look
> them up if you like.
>
> All the best,
>
>
> T
>
> Timothy Feinstein, Ph.D.
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
> Behalf Of Ariel, Pablo
> Sent: Thursday, April 15, 2021 6:05 AM
> To: [hidden email]
> Subject: immersion fluid options for long timelapses with long working
> distance water objective
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
> https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=hA8vGP0sPHELDD2uLakMIMwyzaRT%2FB%2FIzr8W%2BUZTwjQ%3D&amp;reserved=0
> Post images on
> https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=ioe%2BTAGtoI5M2ly81aUixXGRZSSMIMp%2FAUipc0RESoM%3D&amp;reserved=0
> and include the link in your posting.
> *****
>
> Dear all,
>
> I am demoing a long working distance water objective from Leica (HC
> FLUOTAR L 25x/0,95 W VISIR, WD: 2.4mm #506375) in an inverted
> configuration, and am running into some challenges.
>
> While the objective is very nice, the long working distance presents a
> problem for long-duration timelapses with multiple positions. Water works
> fine, but it eventually evaporates (in hours). Zeiss Immersol W does not
> evaporate, but cannot hold the column when the objective is over a mm from
> the coverslip (it drips/slides off). Leica has an automated water
> dispenser, but it is not available for my current demo. I was wondering if
> anyone had any ideas for other immersion options that have a surface
> tension and RI similar to water, but don't evaporate as easily.
>
> Thanks,
> Pablo
>
> Pablo Ariel, Ph.D.
> Assistant Professor
> Director of the Microscopy Services Laboratory Brinkhous-Bullitt Bldg B04
> Department of Pathology and Laboratory Medicine University of North
> Carolina at Chapel Hill
>
> https://nam12.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.med.unc.edu%2Fmicroscopy&amp;data=04%7C01%7Ctnf8%40PITT.EDU%7Cb38ac40b056d4715f22c08d8fff65d75%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637540780890462531%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&amp;sdata=wnOHApiZUAd6ZVqaqBQCrP7O8wCev%2BND97lhM2FmhQY%3D&amp;reserved=0
> Tel: 919-966-2413
>
>
>
>
>
>
>
>
> När du har kontakt med oss på Uppsala universitet med e-post så innebär
> det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör
> det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/
>
> E-mailing Uppsala University means that we will process your personal
> data. For more information on how this is performed, please read here:
> http://www.uu.se/en/about-uu/data-protection-policy
>