Michael Bastiani Michael Bastiani
Reply | Threaded
Open this post in threaded view
♦
♦
|

laser launch protocol

Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
engineer gets back from a training session in Germany). It would be great if I could get it
useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
(two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
someone routinely does this on their own and has step by step instructions I would love to
hear from you.  
thanks
michael
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

If you can hook the other end of the fiber up to some sort of optical
power meter, then you are in business.  Basically stare at the readout
on the meter while adjusting the screws.  Try to scan out a 'raster'
pattern: go back and forth on one axis, while moving the other axis in
a sweep from left to right.  Eventually you will see the power meter
react, which means you are close to the fiber tip.  Then just keep
making small adjustments to optimize for power.  Note that you can get
stuck on local maxima rather than the true maxima.  To prevent this,
when you think you are on the maximum value, move off of it a
significant distance with one axis and see that things drop off, then
move back, then try the other direction.  If you don't see any higher
values anywhere then you are on the true maximum.

Craig


On Thu, Apr 8, 2010 at 4:09 PM, Michael Bastiani
<[hidden email]> wrote:

> Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
> a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
> engineer gets back from a training session in Germany). It would be great if I could get it
> useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
> (two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
> someone routinely does this on their own and has step by step instructions I would love to
> hear from you.
> thanks
> michael
>
Prof Hindsight Prof Hindsight
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

If the blind lead the blind, both shall fall into the ditch. Matthew 15:14 - (laser safety ;-)
 
This is not a trivial process.  To the best of my knowledge, aligning the beam into the fiber optic is not the only step required if the laser has been replaced.  The laser will firstly need retuning/aligning internally and then needs recalibrating.
 
I would suggest contacting your local service organisation and enquire if there is another engineer that can visit sooner.

PH

On 8 April 2010 23:18, Craig Brideau <[hidden email]> wrote:

If you can hook the other end of the fiber up to some sort of optical
power meter, then you are in business.  Basically stare at the readout
on the meter while adjusting the screws.  Try to scan out a 'raster'
pattern: go back and forth on one axis, while moving the other axis in
a sweep from left to right.  Eventually you will see the power meter
react, which means you are close to the fiber tip.  Then just keep
making small adjustments to optimize for power.  Note that you can get
stuck on local maxima rather than the true maxima.  To prevent this,
when you think you are on the maximum value, move off of it a
significant distance with one axis and see that things drop off, then
move back, then try the other direction.  If you don't see any higher
values anywhere then you are on the true maximum.

Craig


On Thu, Apr 8, 2010 at 4:09 PM, Michael Bastiani
<[hidden email]> wrote:
> Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
> a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
> engineer gets back from a training session in Germany). It would be great if I could get it
> useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
> (two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
> someone routinely does this on their own and has step by step instructions I would love to
> hear from you.
> thanks
> michael
>

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

Depending on the complexity of the system it can be risky to do this
sort of alignment.  If you have any experience it is not so bad, but
if you don't know what you are doing and the system is complex you can
make things worse.  Fortunately it seems to have worked out in this
case from what Michael tells me.  I do encourage people to tinker,
otherwise they never learn, but I suggest you go slowly and carefully
and take all necessary safety precautions (goggles,etc.).

Craig


On Fri, Apr 9, 2010 at 6:06 AM, Prof Hindsight
<[hidden email]> wrote:

> If the blind lead the blind, both shall fall into the ditch. Matthew 15:14 -
> (laser safety ;-)
>
> This is not a trivial process.  To the best of my knowledge, aligning the
> beam into the fiber optic is not the only step required if the laser has
> been replaced.  The laser will firstly need retuning/aligning internally and
> then needs recalibrating.
>
> I would suggest contacting your local service organisation and enquire if
> there is another engineer that can visit sooner.
>
> PH
>
> On 8 April 2010 23:18, Craig Brideau <[hidden email]> wrote:
>>
>> If you can hook the other end of the fiber up to some sort of optical
>> power meter, then you are in business.  Basically stare at the readout
>> on the meter while adjusting the screws.  Try to scan out a 'raster'
>> pattern: go back and forth on one axis, while moving the other axis in
>> a sweep from left to right.  Eventually you will see the power meter
>> react, which means you are close to the fiber tip.  Then just keep
>> making small adjustments to optimize for power.  Note that you can get
>> stuck on local maxima rather than the true maxima.  To prevent this,
>> when you think you are on the maximum value, move off of it a
>> significant distance with one axis and see that things drop off, then
>> move back, then try the other direction.  If you don't see any higher
>> values anywhere then you are on the true maximum.
>>
>> Craig
>>
>>
>> On Thu, Apr 8, 2010 at 4:09 PM, Michael Bastiani
>> <[hidden email]> wrote:
>> > Anyone have experience aligning a new KrAr laser on a Microradiance
>> > 2000?  I just installed
>> > a new laser and have a service scheduled in 2-3 weeks to align the laser
>> > (when service
>> > engineer gets back from a training session in Germany). It would be
>> > great if I could get it
>> > useable sooner and I would feel a great sense of accomplishment!   It
>> > looks straightforward
>> > (two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is
>> > not trivial.  If
>> > someone routinely does this on their own and has step by step
>> > instructions I would love to
>> > hear from you.
>> > thanks
>> > michael
>> >
>
>
confocalservice confocalservice
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

In reply to this post by Michael Bastiani
Hi ,its a bit far for me , otherwise I would replace this laser for you free of charge!
 
 
Farhad 
Confocalservice Ltd.
 


 
On 8 April 2010 23:09, Michael Bastiani <[hidden email]> wrote:
Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
engineer gets back from a training session in Germany). It would be great if I could get it
useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
(two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
someone routinely does this on their own and has step by step instructions I would love to
hear from you.
thanks
michael

Alice Rodriguez Diaz Alice Rodriguez Diaz
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

Try  Prof Dan Kiehart, Duke University.

 

Best,

Alice Rodriguez-Diaz

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of ConfocalService Ltd.
Sent: Monday, April 12, 2010 5:50 AM
To: [hidden email]
Subject: Re: laser launch protocol

 

Hi ,its a bit far for me , otherwise I would replace this laser for you free of charge!

 

 

Farhad 

Confocalservice Ltd.

 



 

On 8 April 2010 23:09, Michael Bastiani <[hidden email]> wrote:

Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
engineer gets back from a training session in Germany). It would be great if I could get it
useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
(two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
someone routinely does this on their own and has step by step instructions I would love to
hear from you.
thanks
michael

 

Mike Mathews Mike Mathews
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

Do you have any other lasers on the launch?
 
Mike
 


 
On Mon, Apr 12, 2010 at 10:38 AM, Alice Rodriguez Diaz <[hidden email]> wrote:

Try  Prof Dan Kiehart, Duke University.

 

Best,

Alice Rodriguez-Diaz

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of ConfocalService Ltd.
Sent: Monday, April 12, 2010 5:50 AM
To: [hidden email]
Subject: Re: laser launch protocol

 

Hi ,its a bit far for me , otherwise I would replace this laser for you free of charge!

 

 

Farhad 

Confocalservice Ltd.

 



 

On 8 April 2010 23:09, Michael Bastiani <[hidden email]> wrote:

Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
engineer gets back from a training session in Germany). It would be great if I could get it
useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
(two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
someone routinely does this on their own and has step by step instructions I would love to
hear from you.
thanks
michael

 




--
Mike Mathews
Spectrofuge Corporation
919-361-2197  Office
919-423-1131  Cell
[hidden email]
www.Spectrofuge.com
Alice Rodriguez Diaz Alice Rodriguez Diaz
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: laser launch protocol

Without much more detail of the situation … the tip of the fiber of course cannot be touched, the rest is a matter of extreme patience and very small and delicate movements. It should take about a day or two depending on your setup. But again, send a detailed email to Dan Kiehart (Duke). We did this all the time with Zeiss scopes.

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mike Mathews
Sent: Monday, April 12, 2010 10:17 AM
To: [hidden email]
Subject: Re: laser launch protocol

 

Do you have any other lasers on the launch?

 

Mike

 



 

On Mon, Apr 12, 2010 at 10:38 AM, Alice Rodriguez Diaz <[hidden email]> wrote:

Try  Prof Dan Kiehart, Duke University.

 

Best,

Alice Rodriguez-Diaz

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of ConfocalService Ltd.
Sent: Monday, April 12, 2010 5:50 AM
To: [hidden email]
Subject: Re: laser launch protocol

 

Hi ,its a bit far for me , otherwise I would replace this laser for you free of charge!

 

 

Farhad 

Confocalservice Ltd.

 



 

On 8 April 2010 23:09, Michael Bastiani <[hidden email]> wrote:

Anyone have experience aligning a new KrAr laser on a Microradiance 2000?  I just installed
a new laser and have a service scheduled in 2-3 weeks to align the laser (when service
engineer gets back from a training session in Germany). It would be great if I could get it
useable sooner and I would feel a great sense of accomplishment!   It looks straightforward
(two mirrors, 4 adjustment screws), but I know hitting a 50um fiber is not trivial.  If
someone routinely does this on their own and has step by step instructions I would love to
hear from you.
thanks
michael

 




--