George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear listserv, latest Brainbow paper has some useful imaging advice http://www.nature.com/nmeth/journal/v10/n6/abs/nmeth.2450.html Improved tools for the Brainbow toolbox -pp540 - 547 Dawen Cai, Kimberly B Cohen, Tuanlian Luo, Jeff W Lichtman & Joshua R Sanes doi:10.1038/nmeth.2450 I disagree with a couple of the authors imaging advice on page 544 (most of the advice is very good): Authors: "Fluorophores with overlap in the excitation or emission spectra should be imaged sequentially rather than simultaneously to minimize fluorescence cross-talk and thereby optimize color separation". On a typical confocal microscope (ex. Leica SP5, SP8 or Zeiss LSM710, 780) there are several detectors (ex. 5 on SP5 or SP8) or detector array (ex. LSM710 or 780). Therefore, fluorophores that excite well at a given excitation wavelength should be imaged simultaneously. I also recommend the latest detectors, ex., HyD for leica, GaAsP for Zeiss, photon counting mode if available (and TCSPC lifetime if available). In the "Image processing" paragraph on page 544, the authors suggest (for confocal) "slower scanning or averaging of multiple scans". Slower scanning is more likely to lead to photobleaching (though the authors somewhat mitigate this by recommending low laser power). I recommend resonant mode (confocal or multiphoton), for users who have a resonant scanner. I have previously suggested here on the listserv that acquiring X number of confocal frames without averaging, and then calculating the median for each pixel, is a much better approach than averaging. I encourage the confocal vendors to implement this, or even better "result value selection", based on the noise distribution of the instrument (especially the detectors). Even better would be to acquire the raw (R.S. mode) frames and clean up in the deconvolution algorithm. Speaking of detectors (and tto not leave out cameras): F. Huang, ... J. Bewersdorf have a nice article showing how well sCMOS can work in single molecule localization microscopy ... doi:10.1038/nmeth.2488 current abstract link is http://www.nature.com/nmeth/journal/vaop/ncurrent/abs/nmeth.2488.html A lot of confocal microscope users are still unaware of simple (and free) image processing methods to improve imaging data, such as PiMP, http://jcs.biologists.org/content/125/9/2257.long (I recommend for 63x/1.4 NA confocal, 30 frame acquisition [current plugin is for single plane], filter 1.6, 16-bit output). Contact the authors for the ImageJ plugin. *** The Brainbow 3.x's "author's file" feature is at http://www.nature.com/nmeth/journal/v10/n6/abs/nmeth.2487.html Zebrafish fans and developmental biologists of all hues should find of interest the same group's Zebrabow paper, PMID: 23757414 *** The same issue of Nature Methods has several other microscope imaging articles, including: http://www.nature.com/nmeth/journal/v10/n6/abs/nmeth.2481.html CLARITY for mapping the nervous system http://www.nature.com/nmeth/journal/v10/n6/abs/nmeth.2477.html Mapping brain circuitry with a light microscope Imaging human connectomes at the macroscale http://www.nature.com/nmeth/journal/v10/n6/abs/nmeth.2482.html -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 |
|
Martin Wessendorf-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey, George-- On 6/23/2013 9:56 AM, George McNamara wrote: > I disagree with a couple of the authors imaging advice on page 544 (most > of the advice is very good): > > Authors: "Fluorophores with overlap in the excitation or emission > spectra should be imaged sequentially rather than simultaneously to > minimize fluorescence cross-talk and thereby optimize color separation". > > On a typical confocal microscope (ex. Leica SP5, SP8 or Zeiss LSM710, > 780) there are several detectors (ex. 5 on SP5 or SP8) or detector array > (ex. LSM710 or 780). Therefore, fluorophores that excite well at a given > excitation wavelength should be imaged simultaneously. I also recommend > the latest detectors, ex., HyD for leica, GaAsP for Zeiss, photon > counting mode if available (and TCSPC lifetime if available). I'm confused about why you think this is bad advice. If you have coexpression of labels, it'll be much easier to determine whether or not a particular cell is single-labeled or multiple-labeled if you use a laser that doesn't excite all the candidates (or conversely, if you use a barrier filter that doesn't pass all the candidates). This is particularly true when you have strong expression of one and weak expression of the other. Even with an detector array, there's a limit to what you can ferret out (--unless you have infinite photons, of course!) However, my confusion probably means I'm missing something so explain away! Thanks--take care-- Martin Wessendorf -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
|
George McNamara |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Martin, Sequential acquisition of fluorophores that can be excited at the same wavelength is silly. On a multi-detector microscope like most point scanning confocal microscopes, such as Leica SP5/SP8 (preferably with HyD's) or Zeiss LSM780, the user has 5 (SP#) or up to 34 (LSM710 or 780, 2 standard PMTs, 32 channel spectral detector ... on 780 with GaAsP spectral detector one would probably just use that). To take a simple scenario, with the following three FPs, mTFP1 ("Teal") CY11.5 (Cyan-Venus high FRET; see S.S. Vogel papers for a series of CY fusions with different FRET efficiencies) mBeRFP (446ex, 615em), published in Yang et al 2013 PLoS One http://www.plosone.org/article/info%3Adoi%2F10.1371%2Fjournal.pone.0064849 one could sequentially excite at 458 nm (Argon ion laser), or 440 or 436 nm (other light sources), with single channel acquisition - as I quoted previously. However, all three fluorophores (or even more with Vogel's C-Y series) are going to emit and/or photobleach. So, if you have enough detectors, best to acquire simultaneously. *** One should pay attention to details when imaging, whether point scanning confocal or widefield. For example, Krylova et al 2013 PLoS One http://www.plosone.org/article/info%3Adoi%2F10.1371%2Fjournal.pone.0063286 imaged four FP-FP dimers (in different cells) with single excitation: mCherry-nls-mCherry mRaspberry-nls-mRaspberry mKate2-nls-mKate2 mPlum-nls-mPlum (note: mPlum's are dim, they would have been better off with mPlum-mPlum-mPlum(etc), Steve Vogel has gone up to Venus 6 [V6]). My kudos to them for approximately doubling the brightness of each protein by using tandem dimers (though tdTomato might have been a better choice). Krylova et al used a widefield microscope, 575/15 ex filter (593 dichroic) with 655/40 [635-675nm] and 628/40 [608-648 nm] exciters. Their figure 4 shows the two emission filters overlap in the range of 635-648 nm (bandpass interference filters are not absolutely vertical cut on/off, but close enough). They would have been better with non-overlapping emission bands (and two cameras would have been nice for simultaneous acquisition). An optimal emission filter would be about $150, a lot less than the publication charges for PLoS One or authors time to work on this project. Ms. Vinita Popat, a summer student working in our lab, calculated for me that 600-630 nm for the short wavelenghth emission filter, and 630LP would be the optimal pair (under the simplifying assumption that each FP is equally bright ... as noted above, mPlum could have improved performance with a higher order multimer). *** I recently (re)read a very nice paper from Richard Neher et al, showing advantages to using two or more EXCITATION wavelengths - sequentially - with the same emission bands, to obtain additional information (2x more images, "multiple excitations greatly facilitate the decomposition"). Neher RA, Mitkovski M, Kirchhoff F, Neher E, Theis FJ, Zeug A 2009 Blind source separation techniques for the decomposition of multiply labeled fluorescence images. </pubmed/19413985> Biophys J. 96: 3791-800. doi: 10.1016/j.bpj.2008.10.068. PMID: 19413985 http://www.cell.com/biophysj/retrieve/pii/S0006349509000927 Methods of blind source separation are used in many contexts to separate composite data sets according to their sources. Multiply labeled fluorescence microscopy images represent such sets, in which the sources are the individual labels. Their distributions are the quantities of interest and have to be extracted from the images. This is often challenging, since the recorded emission spectra of fluorescent dyes are environment- and instrument-specific. We have developed a nonnegative matrix factorization (NMF) algorithm to detect and separate spectrally distinct components of multiply labeled fluorescence images. It operates on spectrally resolved images and delivers both the emission spectra of the identified components and images of their abundance. We tested the proposed method using biological samples labeled with up to four spectrally overlapping fluorescent labels. In most cases, NMF accurately decomposed the images into contributions of individual dyes. However, the solutions are not unique when spectra overlap strongly or when images are diffuse in their structure. To arrive at satisfactory results in such cases, we extended NMF to incorporate preexisting qualitative knowledge about spectra and label distributions. We show how data acquired through excitations at two or three different wavelengths can be integrated and that multiple excitations greatly facilitate the decomposition. By allowing reliable decomposition in cases where the spectra of the individual labels are not known or are known only inaccurately, the proposed algorithms greatly extend the range of questions that can be addressed with quantitative microscopy. ImageJ plugins at http://www.mh-hannover.de/cellneurophys/poissonNMF/NMF/ I hope vendors start including this capability. I do wish Prof. Neher had chosen a less mathematically smelly word than decomposition - one reason I prefer to call this spectral unmixing. Enjoy, George On 6/23/2013 2:23 PM, Martin Wessendorf wrote: > Hey, George-- > > On 6/23/2013 9:56 AM, George McNamara wrote: > >> I disagree with a couple of the authors imaging advice on page 544 (most >> of the advice is very good): >> >> Authors: "Fluorophores with overlap in the excitation or emission >> spectra should be imaged sequentially rather than simultaneously to >> minimize fluorescence cross-talk and thereby optimize color separation". >> >> On a typical confocal microscope (ex. Leica SP5, SP8 or Zeiss LSM710, >> 780) there are several detectors (ex. 5 on SP5 or SP8) or detector array >> (ex. LSM710 or 780). Therefore, fluorophores that excite well at a given >> excitation wavelength should be imaged simultaneously. I also recommend >> the latest detectors, ex., HyD for leica, GaAsP for Zeiss, photon >> counting mode if available (and TCSPC lifetime if available). > > I'm confused about why you think this is bad advice. If you have > coexpression of labels, it'll be much easier to determine whether or > not a particular cell is single-labeled or multiple-labeled if you use > a laser that doesn't excite all the candidates (or conversely, if you > use a barrier filter that doesn't pass all the candidates). This is > particularly true when you have strong expression of one and weak > expression of the other. Even with an detector array, there's a limit > to what you can ferret out (--unless you have infinite photons, of > course!) > > However, my confusion probably means I'm missing something so explain > away! > > Thanks--take care-- > > Martin Wessendorf > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 |
| Free forum by Nabble | Edit this page |