mCherry and 2P

classic Classic list List threaded Threaded
♦
♦
7 messages Options
Koo, Lily (NIH/NIAID) [E] Koo, Lily (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

mCherry and 2P

Dear Listers,

I am wondering if any of you have had success imaging mCherry in the 2-photon mode?  We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P.  We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success.  Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore.  

Any suggestions?

Thanks,

Lily
Quoc Thang Nguyen Quoc Thang Nguyen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

> Dear Listers,
>
> I am wondering if any of you have had success imaging mCherry in the
> 2-photon mode?  We have mCherry bacteria rods that are very bright in
> confocal, but cannot get an image in 2P.  We tried up to 1030nm (the
> coherent laser stops at 1040 and the power drops down to ~600mW at 1030),
> but with no success.  Around 900nm we could detect a shadowy image of the
> bacteria, but at a power level that immediately bleaches the fluorophore.
>
> Any suggestions?
>
> Thanks,
>
> Lily
>

Try 740-760 nm. Works great.
Greg Martin-8 Greg Martin-8
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

In reply to this post by Koo, Lily (NIH/NIAID) [E]
Hey Lily --
 
    I just did some 2-photon imaging of mCherry at 1000nm and it worked great with little bleaching.  This was on fixed mammalian cells with the Spect. Phys. HP deep see laser with 15% transmission.  Sorry, don't have a real number for power for comparison.  Do you have non-descanned detectors on your system?  Have you been able to image any other red fluors with your 2-photon? 
 
Be peace!  Greg.
 
Greg Martin
 
Keck Microscopy Facility
University of Washington
www.depts.washington.edu/keck
 
Lynn and Mike Garvey Cell Imaging Laboratory
University of Washington Institute for Stem Cell and Regenerative Medicine
 
206-685-8784 (office)
425-344-2632 (cell)
----- Original Message -----
From: "Koo, Lily (NIH/NIAID) [E]" <[hidden email]>
Sent: Tuesday, June 23, 2009 11:26 AM
Subject: mCherry and 2P

Dear Listers,

I am wondering if any of you have had success imaging mCherry in the 2-photon mode?  We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P.  We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success.  Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. 

Any suggestions?

Thanks,

Lily
Rainer Kohler-2 Rainer Kohler-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

In reply to this post by Koo, Lily (NIH/NIAID) [E]
Hi Lily,

I recently imaged an mCherry fusion protein  in Hela cells with a Mai
Tai DeepSee laser on an Olympus Fv1000 system. It had a maximum
signal at around 760  nm.

Best regards

Rainer


>Dear Listers,
>
>I am wondering if any of you have had success imaging mCherry in the
>2-photon mode?  We have mCherry bacteria rods that are very bright
>in confocal, but cannot get an image in 2P.  We tried up to 1030nm
>(the coherent laser stops at 1040 and the power drops down to ~600mW
>at 1030), but with no success.  Around 900nm we could detect a
>shadowy image of the bacteria, but at a power level that immediately
>bleaches the fluorophore.
>
>Any suggestions?
>
>Thanks,
>
>Lily


--
Rainer Kohler, Ph.D
Assistant in Research
Cell Phone: 978 578 5057
E.mail: [hidden email]

Center for Systems Biology
Massachusetts General Hospital
Richard B. Simches Research Center
185 Cambridge Street
Suite 5.210
Boston, MA 02114
Phone: (617) 643-0500



The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

In reply to this post by Koo, Lily (NIH/NIAID) [E]
I've never tried this but looking at all the other replies my
guess would be that your IR blocking filter in the MP is blocking
the mCherry fluorescence.  Check out what that filter is.  I certainly
find that our blocking filter on our Leica SP2 MP blocks chlorophyll
fluorescence, for example.  What is your blocking filter?

                                       Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E]
Sent: Wednesday, 24 June 2009 4:26 AM
To: [hidden email]
Subject: mCherry and 2P

Dear Listers,

I am wondering if any of you have had success imaging mCherry in the 2-photon mode?  We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P.  We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success.  Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore.  

Any suggestions?

Thanks,

Lily

Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM
 

Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM
 
Koo, Lily (NIH/NIAID) [E] Koo, Lily (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

Thanks for all the replies.  I am working on our collaborator's Meta510 and just getting warmed up to the system.  They do have a non-descanned detector, which didn't help us in capturing mCherry signal when we tried from 800-1030.  At 800, we started to see some signal, but there was significant bleaching.  I believe the blocking filter is set at 690nm.  We will try 740nm ex. tomorrow and will let you all know how it will work out.

Thanks,

Lily

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox [[hidden email]]
Sent: Tuesday, June 23, 2009 9:41 PM
To: [hidden email]
Subject: Re: mCherry and 2P

I've never tried this but looking at all the other replies my
guess would be that your IR blocking filter in the MP is blocking
the mCherry fluorescence.  Check out what that filter is.  I certainly
find that our blocking filter on our Leica SP2 MP blocks chlorophyll
fluorescence, for example.  What is your blocking filter?

                                       Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E]
Sent: Wednesday, 24 June 2009 4:26 AM
To: [hidden email]
Subject: mCherry and 2P

Dear Listers,

I am wondering if any of you have had success imaging mCherry in the 2-photon mode?  We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P.  We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success.  Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore.

Any suggestions?

Thanks,

Lily

Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM


Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM
Koo, Lily (NIH/NIAID) [E] Koo, Lily (NIH/NIAID) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mCherry and 2P

Dear all,

I would like to inform the list that we had good signal on mCherry when we excited it around 740nm.  The signal dropped significantly after 790 (we hope to excite both mCherry and GFP).  Our prior lack of success was apparently due to the fact that we were trying it on fixed samples (mCherry labelled bacteria), which killed the emission nearly entirely.  

Thanks,

Lily

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E]
Sent: Wednesday, June 24, 2009 1:23 AM
To: [hidden email]
Subject: Re: mCherry and 2P

Thanks for all the replies.  I am working on our collaborator's Meta510 and just getting warmed up to the system.  They do have a non-descanned detector, which didn't help us in capturing mCherry signal when we tried from 800-1030.  At 800, we started to see some signal, but there was significant bleaching.  I believe the blocking filter is set at 690nm.  We will try 740nm ex. tomorrow and will let you all know how it will work out.

Thanks,

Lily

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox [[hidden email]]
Sent: Tuesday, June 23, 2009 9:41 PM
To: [hidden email]
Subject: Re: mCherry and 2P

I've never tried this but looking at all the other replies my
guess would be that your IR blocking filter in the MP is blocking
the mCherry fluorescence.  Check out what that filter is.  I certainly
find that our blocking filter on our Leica SP2 MP blocks chlorophyll
fluorescence, for example.  What is your blocking filter?

                                       Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E]
Sent: Wednesday, 24 June 2009 4:26 AM
To: [hidden email]
Subject: mCherry and 2P

Dear Listers,

I am wondering if any of you have had success imaging mCherry in the 2-photon mode?  We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P.  We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success.  Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore.

Any suggestions?

Thanks,

Lily

Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM


Internal Virus Database is out-of-date.
Checked by AVG.
Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM