Dear Listers,
I am wondering if any of you have had success imaging mCherry in the 2-photon mode? We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P. We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success. Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. Any suggestions? Thanks, Lily |
|
|
> Dear Listers,
> > I am wondering if any of you have had success imaging mCherry in the > 2-photon mode? We have mCherry bacteria rods that are very bright in > confocal, but cannot get an image in 2P. We tried up to 1030nm (the > coherent laser stops at 1040 and the power drops down to ~600mW at 1030), > but with no success. Around 900nm we could detect a shadowy image of the > bacteria, but at a power level that immediately bleaches the fluorophore. > > Any suggestions? > > Thanks, > > Lily > Try 740-760 nm. Works great. |
|
|
In reply to this post by Koo, Lily (NIH/NIAID) [E]
Hey Lily --
I just did some 2-photon
imaging of mCherry at 1000nm and it worked great with little bleaching.
This was on fixed mammalian cells with the Spect. Phys. HP deep see
laser with 15% transmission. Sorry, don't have a real number for
power for comparison. Do you have non-descanned detectors on your
system? Have you been able to image any other red fluors with your
2-photon?
Be peace! Greg.
Greg Martin
Lynn and Mike Garvey Cell Imaging
Laboratory
University of Washington Institute for Stem Cell and Regenerative Medicine 206-685-8784 (office)
425-344-2632 (cell) ----- Original Message -----
From: "Koo, Lily (NIH/NIAID) [E]"
<[hidden email]>
To: <[hidden email]>
Sent: Tuesday, June 23, 2009 11:26
AM
Subject: mCherry and 2P I am wondering if any of you have had success imaging mCherry in the 2-photon mode? We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P. We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success. Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. Any suggestions? Thanks, Lily |
|
|
In reply to this post by Koo, Lily (NIH/NIAID) [E]
Hi Lily,
I recently imaged an mCherry fusion protein in Hela cells with a Mai Tai DeepSee laser on an Olympus Fv1000 system. It had a maximum signal at around 760 nm. Best regards Rainer >Dear Listers, > >I am wondering if any of you have had success imaging mCherry in the >2-photon mode? We have mCherry bacteria rods that are very bright >in confocal, but cannot get an image in 2P. We tried up to 1030nm >(the coherent laser stops at 1040 and the power drops down to ~600mW >at 1030), but with no success. Around 900nm we could detect a >shadowy image of the bacteria, but at a power level that immediately >bleaches the fluorophore. > >Any suggestions? > >Thanks, > >Lily -- Rainer Kohler, Ph.D Assistant in Research Cell Phone: 978 578 5057 E.mail: [hidden email] Center for Systems Biology Massachusetts General Hospital Richard B. Simches Research Center 185 Cambridge Street Suite 5.210 Boston, MA 02114 Phone: (617) 643-0500 The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
|
|
In reply to this post by Koo, Lily (NIH/NIAID) [E]
I've never tried this but looking at all the other replies my
guess would be that your IR blocking filter in the MP is blocking the mCherry fluorescence. Check out what that filter is. I certainly find that our blocking filter on our Leica SP2 MP blocks chlorophyll fluorescence, for example. What is your blocking filter? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E] Sent: Wednesday, 24 June 2009 4:26 AM To: [hidden email] Subject: mCherry and 2P Dear Listers, I am wondering if any of you have had success imaging mCherry in the 2-photon mode? We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P. We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success. Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. Any suggestions? Thanks, Lily Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM |
|
|
Thanks for all the replies. I am working on our collaborator's Meta510 and just getting warmed up to the system. They do have a non-descanned detector, which didn't help us in capturing mCherry signal when we tried from 800-1030. At 800, we started to see some signal, but there was significant bleaching. I believe the blocking filter is set at 690nm. We will try 740nm ex. tomorrow and will let you all know how it will work out.
Thanks, Lily ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox [[hidden email]] Sent: Tuesday, June 23, 2009 9:41 PM To: [hidden email] Subject: Re: mCherry and 2P I've never tried this but looking at all the other replies my guess would be that your IR blocking filter in the MP is blocking the mCherry fluorescence. Check out what that filter is. I certainly find that our blocking filter on our Leica SP2 MP blocks chlorophyll fluorescence, for example. What is your blocking filter? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E] Sent: Wednesday, 24 June 2009 4:26 AM To: [hidden email] Subject: mCherry and 2P Dear Listers, I am wondering if any of you have had success imaging mCherry in the 2-photon mode? We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P. We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success. Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. Any suggestions? Thanks, Lily Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM |
|
|
Dear all,
I would like to inform the list that we had good signal on mCherry when we excited it around 740nm. The signal dropped significantly after 790 (we hope to excite both mCherry and GFP). Our prior lack of success was apparently due to the fact that we were trying it on fixed samples (mCherry labelled bacteria), which killed the emission nearly entirely. Thanks, Lily ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E] Sent: Wednesday, June 24, 2009 1:23 AM To: [hidden email] Subject: Re: mCherry and 2P Thanks for all the replies. I am working on our collaborator's Meta510 and just getting warmed up to the system. They do have a non-descanned detector, which didn't help us in capturing mCherry signal when we tried from 800-1030. At 800, we started to see some signal, but there was significant bleaching. I believe the blocking filter is set at 690nm. We will try 740nm ex. tomorrow and will let you all know how it will work out. Thanks, Lily ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox [[hidden email]] Sent: Tuesday, June 23, 2009 9:41 PM To: [hidden email] Subject: Re: mCherry and 2P I've never tried this but looking at all the other replies my guess would be that your IR blocking filter in the MP is blocking the mCherry fluorescence. Check out what that filter is. I certainly find that our blocking filter on our Leica SP2 MP blocks chlorophyll fluorescence, for example. What is your blocking filter? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Koo, Lily (NIH/NIAID) [E] Sent: Wednesday, 24 June 2009 4:26 AM To: [hidden email] Subject: mCherry and 2P Dear Listers, I am wondering if any of you have had success imaging mCherry in the 2-photon mode? We have mCherry bacteria rods that are very bright in confocal, but cannot get an image in 2P. We tried up to 1030nm (the coherent laser stops at 1040 and the power drops down to ~600mW at 1030), but with no success. Around 900nm we could detect a shadowy image of the bacteria, but at a power level that immediately bleaches the fluorophore. Any suggestions? Thanks, Lily Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM Internal Virus Database is out-of-date. Checked by AVG. Version: 7.5.560 / Virus Database: 270.12.26/2116 - Release Date: 15/05/2009 6:16 AM |