![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hey I hope somebody can help me out. I've started working with a fusion protein of mEos4a and the photoconversion has been working poorly. I'm using a swept field confocal and Prairie View software. I am using a 488 laser at 20% power and 561 at 35% power for visualizing the protein in live cells. I notice that the green form of mEos bleaches quite quickly even as I'm identifying cells of interest to image. In addition, there tends to be quite a bit of red signal already detectable in these cells without photoconversion by a 405 laser. I am able to get ~500% increase in red signal and maybe a 50% decrease in green signal upon 405 conversion (but still far from complete and it was difficult to find cells with minimal red signal). I converted using 20 fast repetitions of a relatively low powered 405 (~10%) at a single point (0.500micron diameter). Does anyone have any advice for best visualization laser and photoconversion laser settings? I'm looking to use mEos4 for FRAP/FLAP. Is mEos sensitive to ambient light Best, Jordan |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Jordan, i am working with mEos2 but dont have experience with mEos4, but still i would like to suggest two things: 1) i have observed photoconversion to some extent by illumination with a UV-Lamp and a green filter when the lamp is very strong. Is it possible to identify the cells in brightfield and image in the red channel first to see whether the 488nm illumination already photoconverts? 2) regarding the photoconversion for me it works better to have a high power output and less repetitions, but i dont know if thats genereally the case. In the beginning i tried to photoconvert the total green signal to red aswell, but realised that i dont yield a strong signal at the end anymore most likely due to bleaching caused by high energy input. Additionally, at least my cells, would not survive this procedure. Now i´m working with approximately 50/50 green to red signal. Best, Sarah |
![]() |
In reply to this post by Jordan Becker
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jordan, I'll be around Bill's lab next week a few times collecting data. In line with Sarah's suggestion, I can show you how to use the transmitted light along with DIC in order to find your cells. Jimmy and I talked about this last night after you left. It's a new capability where the system can do DIC imaging through the confocal scanner. Best Jeff Stuckey -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jordan Becker Sent: Friday, May 29, 2015 8:35 AM To: [hidden email] Subject: mEos4 photoconversion ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hey I hope somebody can help me out. I've started working with a fusion protein of mEos4a and the photoconversion has been working poorly. I'm using a swept field confocal and Prairie View software. I am using a 488 laser at 20% power and 561 at 35% power for visualizing the protein in live cells. I notice that the green form of mEos bleaches quite quickly even as I'm identifying cells of interest to image. In addition, there tends to be quite a bit of red signal already detectable in these cells without photoconversion by a 405 laser. I am able to get ~500% increase in red signal and maybe a 50% decrease in green signal upon 405 conversion (but still far from complete and it was difficult to find cells with minimal red signal). I converted using 20 fast repetitions of a relatively low powered 405 (~10%) at a single point (0.500micron diameter). Does anyone have any advice for best visualization laser and photoconversion laser settings? I'm looking to use mEos4 for FRAP/FLAP. Is mEos sensitive to ambient light Best, Jordan |
![]() |
In reply to this post by Jordan Becker
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Jeff, It's great to find you on here! I wondered if the LED was sending some shorter wavelength light that was prematurely switching some to red. From what I can find in previous literature, low power on the 488 laser and high on the 561 is a good starting place for visualizing. I think if I can decrease the premature red signal, the photoconversion ratio should improve accordingly. Thanks for the help Sarah and I'll hope to see you around next week Jeff. Best, Jordan |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jordan We weren't using the transmitted LED the other day. For visualizing with transmitted light we could put a filter in the path that would block the absorption wavelengths for eos if that was a concern. Best Jeff Sent from my iPhone > On May 30, 2015, at 10:15 AM, "Jordan Becker" <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Jeff, > > It's great to find you on here! I wondered if the LED was sending some > shorter wavelength light that was prematurely switching some to red. From > what I can find in previous literature, low power on the 488 laser and high > on the 561 is a good starting place for visualizing. I think if I can > decrease the premature red signal, the photoconversion ratio should improve > accordingly. > > Thanks for the help Sarah and I'll hope to see you around next week Jeff. > > Best, > Jordan |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Early last week somebody brought us mEos to image by TIRF. Even with the 488 nm laser low with ND filters in the path and the EMCCD gain all the way up, we could not image more than 2 or 3 seconds. (It was TIRF, so I guess we effectively showed that there isn't transport of new protein from above down to the substratum, but this wasn't the experiment...) However, when we set the 405 laser low and the 561 laser low and used both or a second of 405 alone followed by continuous 561, we were able to image continuously in the red for 1-2 minutes. So imaging green nearly impossible, Imaging the photoconverted in red, easy. _________________________________________ Michael Cammer, Optical Microscopy Specialist http://ocs.med.nyu.edu/microscopy http://microscopynotes.com/ Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Stuckey, Jeff [[hidden email]] Sent: Saturday, May 30, 2015 11:40 AM To: [hidden email] Subject: Re: mEos4 photoconversion ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jordan We weren't using the transmitted LED the other day. For visualizing with transmitted light we could put a filter in the path that would block the absorption wavelengths for eos if that was a concern. Best Jeff Sent from my iPhone > On May 30, 2015, at 10:15 AM, "Jordan Becker" <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Jeff, > > It's great to find you on here! I wondered if the LED was sending some > shorter wavelength light that was prematurely switching some to red. From > what I can find in previous literature, low power on the 488 laser and high > on the 561 is a good starting place for visualizing. I think if I can > decrease the premature red signal, the photoconversion ratio should improve > accordingly. > > Thanks for the help Sarah and I'll hope to see you around next week Jeff. > > Best, > Jordan |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Early last week somebody brought us mEos to image by TIRF (first time for us, disclaimer: n=1). Even with the 488 nm laser low with ND filters in the path and the EMCCD gain all the way up, we could not image more than 2 or 3 seconds. (It was TIRF, so I guess we effectively showed that there isn't transport of new protein from above down to the substratum, but this wasn't the experiment...) However, when we set the 405 laser low and the 561 laser low and used both or a second of 405 alone followed by continuous 561, we were able to image continuously in the red for 1-2 minutes. So imaging green nearly impossible, Imaging the photoconverted in red, easy. _________________________________________ Michael Cammer, Optical Microscopy Specialist http://ocs.med.nyu.edu/microscopy http://microscopynotes.com/ Cell: (914) 309-3270 ________________________________________ From: Cammer, Michael Sent: Sunday, May 31, 2015 6:28 PM To: Confocal Microscopy List Subject: RE: mEos4 photoconversion Early last week somebody brought us mEos to image by TIRF. Even with the 488 nm laser low with ND filters in the path and the EMCCD gain all the way up, we could not image more than 2 or 3 seconds. (It was TIRF, so I guess we effectively showed that there isn't transport of new protein from above down to the substratum, but this wasn't the experiment...) However, when we set the 405 laser low and the 561 laser low and used both or a second of 405 alone followed by continuous 561, we were able to image continuously in the red for 1-2 minutes. So imaging green nearly impossible, Imaging the photoconverted in red, easy. _________________________________________ Michael Cammer, Optical Microscopy Specialist http://ocs.med.nyu.edu/microscopy http://microscopynotes.com/ Cell: (914) 309-3270 ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Stuckey, Jeff [[hidden email]] Sent: Saturday, May 30, 2015 11:40 AM To: [hidden email] Subject: Re: mEos4 photoconversion ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jordan We weren't using the transmitted LED the other day. For visualizing with transmitted light we could put a filter in the path that would block the absorption wavelengths for eos if that was a concern. Best Jeff Sent from my iPhone > On May 30, 2015, at 10:15 AM, "Jordan Becker" <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Jeff, > > It's great to find you on here! I wondered if the LED was sending some > shorter wavelength light that was prematurely switching some to red. From > what I can find in previous literature, low power on the 488 laser and high > on the 561 is a good starting place for visualizing. I think if I can > decrease the premature red signal, the photoconversion ratio should improve > accordingly. > > Thanks for the help Sarah and I'll hope to see you around next week Jeff. > > Best, > Jordan |
![]() |
In reply to this post by Stuckey, Jeff
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hello list, Please share your experience and advice on CCD camera for imaging fixed slides on widefield fluorescent microscope. We are looking for a reasonably priced, reliable and sensitive camera. High speed is not very important. Also how newest sCMOS compare to CCD? Our 5 year old 1.4MP cooled CCD camera died, and as it is discontinued, we prefer to replace it instead of repairing. It has been mostly used for imaging fixed brain slices labeled by immunofluorescence or fluorescent proteins, some labels being quite dim. Thanks, Arvydas Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core SUNY Upstate Medical University Neuroscience & Physiology Dept |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Arvydas, The sCMOS cameras from (alphabetical order) Andor, Hamamatsu and PCO have similar pixel size but 5x or 5x as many pixels as your old CCD, so a lot more area covered. The sCMOS focusing is a lot faster: 100 fps full frame ... the lower price models are 30 fps (ex. USB3), still faster than an old CCD. If you have limited budget, go with an A,H, or P sCMOS that is USB3 (lower price, still 16-bit ... play with contrast) and invest in LED illuminator, maybe for price consider ThorLabs 4LED https://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=3836 new filter(s), -- ideally a matched Quad filter set --- and find more money for a Sutter + Semrock fast emission wheel with threaded filters - similar cost to slow wheels, but 'time is money". http://www.semrock.com/how-to-get-the-most-out-of-your-sutter-filter-wheel.aspx the right LEDs, right Quad set means no moving parts on the excitation side. Emission filter wheel (especially assuming wide emission bands on the Quad set) would enable you to choose what fluorophore(s). If you can find additional money (neighboring labs? friendly dean?) maybe time to get at least one new objective lens or lenses. Finally, "instant gratification" deconvolution is just about here -- http://www.microvolution.com/ and http://www.microvolution.com/technology.php (yes, my image is on their home page - my colleague went "Wow" when he saw the deconvolved data), along with NVidia TITAN X card or cards (hey, at least you wouldn't need to buy a new PC) ... new GPU cards would also let you put HD 4K monitors on your system. Finally, an data center quality solid state drive for local file saving (and then transfer to the network to a server for long term storage and other desktop PC's access). Do me a favor - for any experiment, do not use overlapping emission filters like this paper did http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3653935/figure/pone-0063286-g004/ the emission filters are about $125 each (a bit more for threaded) -- a tiny fraction of PLoS One's page charges, to say nothing of the money spent on salaries for that paper (they also should have multimerized the mPlum and other dim FPs). If you do like the above paper, see 26010570 for new (one week) paper that cites some of the recent developments in many color cells ('brainbow 2015'). you may also find of use, http://works.bepress.com/gmcnamara/65/ and my other writing on Tattletales and T-Bow. enjoy, George On 6/2/2015 10:12 AM, Arvydas Matiukas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hello list, > > Please share your experience and advice on CCD camera > for imaging fixed slides on widefield fluorescent microscope. > We are looking for a reasonably priced, reliable and sensitive > camera. High speed is not very important. Also how newest sCMOS > compare to CCD? > > Our 5 year old 1.4MP cooled CCD camera died, and as it > is discontinued, we prefer to replace it instead of repairing. > It has been mostly used for imaging fixed brain slices > labeled by immunofluorescence or fluorescent proteins, > some labels being quite dim. > > Thanks, > Arvydas > > Arvydas Matiukas, Ph.D. > Director of Confocal&Two-Photon Core > SUNY Upstate Medical University > Neuroscience& Physiology Dept > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/42 |
![]() |
In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I have imaged mEos2 in live cells, using a Zeiss 700 confocal with no problems at all. It is important to keep laser intensities low of course as Eos does bleach quite quickly in the green. Meanwhile - no fluorescence in the red channel. I have also photo activated mEos2 using the 405nm laser, and measured changes to red-fluorescence (an inverted FRAP experiment) - (using quite low levels of blue light, and not many interations, - set up as a FRAP experiment) A bit of fiddling to get it to work, but it did work well in the end. Can send images to anyone who is interested. All best Michelle On 31/05/2015 23:28, "Cammer, Michael" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Early last week somebody brought us mEos to image by TIRF. Even with the >488 nm laser low with ND filters in the path and the EMCCD gain all the >way up, we could not image more than 2 or 3 seconds. (It was TIRF, so I >guess we effectively showed that there isn't transport of new protein >from above down to the substratum, but this wasn't the experiment...) > >However, when we set the 405 laser low and the 561 laser low and used >both or a second of 405 alone followed by continuous 561, we were able to >image continuously in the red for 1-2 minutes. > >So imaging green nearly impossible, Imaging the photoconverted in red, >easy. > >_________________________________________ >Michael Cammer, Optical Microscopy Specialist >http://ocs.med.nyu.edu/microscopy >http://microscopynotes.com/ >Cell: (914) 309-3270 > >________________________________________ >From: Confocal Microscopy List [[hidden email]] on >behalf of Stuckey, Jeff [[hidden email]] >Sent: Saturday, May 30, 2015 11:40 AM >To: [hidden email] >Subject: Re: mEos4 photoconversion > >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >Hi Jordan > >We weren't using the transmitted LED the other day. For visualizing with >transmitted light we could put a filter in the path that would block the >absorption wavelengths for eos if that was a concern. > >Best > >Jeff > >Sent from my iPhone > >> On May 30, 2015, at 10:15 AM, "Jordan Becker" <[hidden email]> >>wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >>posting. >> ***** >> >> Jeff, >> >> It's great to find you on here! I wondered if the LED was sending some >> shorter wavelength light that was prematurely switching some to red. >>From >> what I can find in previous literature, low power on the 488 laser and >>high >> on the 561 is a good starting place for visualizing. I think if I can >> decrease the premature red signal, the photoconversion ratio should >>improve >> accordingly. >> >> Thanks for the help Sarah and I'll hope to see you around next week >>Jeff. >> >> Best, >> Jordan |
Free forum by Nabble | Edit this page |