Hi
This is probably a very foolish and basic question, but I am going to ask
it any way and just apologise for my ignorance as a biological specialist
using a confocal microscope.

In microscopy I would generally multiply the objective lens magnification
by the ocular lens (often 10x) to get the total magnification of an image.
I want to know if this also applies for the confocal microscope or is the
light path such that the objective magnification is the only one relevant?
If I do need to include additional magnification what would these
magnifications be, are they dependent upon the system being used (LSM 510
META)  or is there a standard magnification?

Thanks for your help in clarifying this for me.

Regards
Nicola
Arne Seitz Arne Seitz
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi,

magnification can be very dangerous parameter. You obtain a magnification of 1 per definition only if you look at an object at the conventional viewing distance of 25 cm. Thus either changing the viewing distance or changing the size of an electronic image will change the magnification of the image. Therefore it is much better to include a scale bar in the image. This automatically also changes if you change the viewing distance and or the size of the image.

In order to create a scale for an image of an wide-field microscope you need to know the total magnification of you microscope (=magnification of the objective fold intermediate magnification; latter on can also be 1) and the pixel size of your camera (if you use binning on the camera you also have to account for this). If you are unsure of the total magnification and/or the pixel size you can also use a calibration slide to measure it (see following link:
http://www.swarthmore.edu/NatSci/nkaplin1/scalebar.htm).

For confocal microscopy it is a little more complicated as you have various other parameters which influence the pixel size in an image (e.g. zoom factor, image size). Therefore all of the images of commercially available confocal microscopes save this information with the image (sometimes maybe not for all of the image formats). So you do not have to care about it. If you open/imort the images with ImageJ (with the LOCI Bioformats plugin you can read even the proprietary company formats http://loci.wisc.edu/software/bio-formats) you can directly add a scale bar via: Analyze-->Tools-->Scale Bar...

Best regards
Arne
 

> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of Nicola
> Green
> Sent: vendredi 16 août 2013 12:57
> To: [hidden email]
> Subject: magnification
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi
> This is probably a very foolish and basic question, but I am going to ask it any
> way and just apologise for my ignorance as a biological specialist using a
> confocal microscope.
>
> In microscopy I would generally multiply the objective lens magnification by
> the ocular lens (often 10x) to get the total magnification of an image.
> I want to know if this also applies for the confocal microscope or is the light
> path such that the objective magnification is the only one relevant?
> If I do need to include additional magnification what would these
> magnifications be, are they dependent upon the system being used (LSM
> 510
> META)  or is there a standard magnification?
>
> Thanks for your help in clarifying this for me.
>
> Regards
> Nicola
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

In reply to this post by Nicola Green
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

In a confocal image the magnification depends on the objective magnification and the area scanned on the sample.  There is of course an 'ocular' or transfer lens but that is not variable and thus is included in the calibration of the system.  So when you 'zoom' in a confocal image you are just scanning a smaller part of the field of view.  The system should still always give you the correct magnification, but since this is a digital image the appropriate units are pixels per micrometre (or the converse, the size of a pixel).  Every confocal microscope will give you these figures.  You should check once in a while, of course, using a standard specimen such as a stage micrometer.  

                                                        Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Nicola Green
Sent: Friday, 16 August 2013 8:57 PM
To: [hidden email]
Subject: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi
This is probably a very foolish and basic question, but I am going to ask
it any way and just apologise for my ignorance as a biological specialist
using a confocal microscope.

In microscopy I would generally multiply the objective lens magnification
by the ocular lens (often 10x) to get the total magnification of an image.
I want to know if this also applies for the confocal microscope or is the
light path such that the objective magnification is the only one relevant?
If I do need to include additional magnification what would these
magnifications be, are they dependent upon the system being used (LSM 510
META)  or is there a standard magnification?

Thanks for your help in clarifying this for me.

Regards
Nicola
JOEL B. SHEFFIELD JOEL B. SHEFFIELD
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

This is actually an issue that befuddles many microscopists.
Magnification, after all, really depends on the final image.  That is, an
object viewed with a 10X lens and viewed through the eyepiece could be said
to have a certain magnification as compared to an object seen with the
unaided eye at a given distance.  However, when that image is captured with
a digital, or photographic system, and displayed on a 10" monitor, or
projected on a 30' screen, the magnification is clearly different. A
printed image, might be yet another size.   It should be standard procedure
to produce images with an embedded size indication, rather than give the
magnification.  The important information is provided by specifying the
objective lens (ideally), which makes the major contribution to
resolution,   So, to get back to the original question, and echo Guy's
comment, , what matters are the dimensions of the area captured by the
system. Most confocal microscopes, as Guy mentions, include this kind of
information about each image, and can be extracted by software.

Joel



On Fri, Aug 16, 2013 at 7:59 AM, Guy Cox <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> In a confocal image the magnification depends on the objective
> magnification and the area scanned on the sample.  There is of course an
> 'ocular' or transfer lens but that is not variable and thus is included in
> the calibration of the system.  So when you 'zoom' in a confocal image you
> are just scanning a smaller part of the field of view.  The system should
> still always give you the correct magnification, but since this is a
> digital image the appropriate units are pixels per micrometre (or the
> converse, the size of a pixel).  Every confocal microscope will give you
> these figures.  You should check once in a while, of course, using a
> standard specimen such as a stage micrometer.
>
>                                                         Guy
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of Nicola Green
> Sent: Friday, 16 August 2013 8:57 PM
> To: [hidden email]
> Subject: magnification
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi
> This is probably a very foolish and basic question, but I am going to ask
> it any way and just apologise for my ignorance as a biological specialist
> using a confocal microscope.
>
> In microscopy I would generally multiply the objective lens magnification
> by the ocular lens (often 10x) to get the total magnification of an image.
> I want to know if this also applies for the confocal microscope or is the
> light path such that the objective magnification is the only one relevant?
> If I do need to include additional magnification what would these
> magnifications be, are they dependent upon the system being used (LSM 510
> META)  or is there a standard magnification?
>
> Thanks for your help in clarifying this for me.
>
> Regards
> Nicola
>



--


Joel B. Sheffield, Ph.D
Department of Biology
Temple University
Philadelphia, PA 19122
Voice: 215 204 8839
e-mail: [hidden email]
URL:  http://astro.temple.edu/~jbs
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

In reply to this post by Guy Cox-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

So how does the manufacturer know the size of the pixel? I guess this will be calibrated at setup of the system with a target of known size for one objective and then stored somewhere in the parameters and hopfully regularly checked. Do they then just change it proportionally for the other objectives? Sometimes objective magnification can be a little bit different than what is written on the barrel, epecially when the lens has a correction collar, did anyone check this?

We had the case that a journal insisted on us giving the magnification in the figure caption, so we gave the objective magnification and insisted on having the scale bar in addition to that.

Andreas

 

 

 

-----Original Message-----
From: Guy Cox <[hidden email]>
To: CONFOCALMICROSCOPY <[hidden email]>
Sent: Fri, 16 Aug 2013 13:01
Subject: Re: magnification


*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

In a confocal image the magnification depends on the objective magnification and
the area scanned on the sample.  There is of course an 'ocular' or transfer lens
but that is not variable and thus is included in the calibration of the system.  
So when you 'zoom' in a confocal image you are just scanning a smaller part of
the field of view.  The system should still always give you the correct
magnification, but since this is a digital image the appropriate units are
pixels per micrometre (or the converse, the size of a pixel).  Every confocal
microscope will give you these figures.  You should check once in a while, of
course, using a standard specimen such as a stage micrometer.  

                                                        Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Nicola Green
Sent: Friday, 16 August 2013 8:57 PM
To: [hidden email]
Subject: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi
This is probably a very foolish and basic question, but I am going to ask
it any way and just apologise for my ignorance as a biological specialist
using a confocal microscope.

In microscopy I would generally multiply the objective lens magnification
by the ocular lens (often 10x) to get the total magnification of an image.
I want to know if this also applies for the confocal microscope or is the
light path such that the objective magnification is the only one relevant?
If I do need to include additional magnification what would these
magnifications be, are they dependent upon the system being used (LSM 510
META)  or is there a standard magnification?

Thanks for your help in clarifying this for me.

Regards
Nicola

 
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I agree.  With a confocal image the scale bar is really the only useful
descriptor.  'Magnification' for confocals is more a function of the timing
of the acquisition electronics, and talking about it in ocular terms is
misleading.

Craig


On Fri, Aug 16, 2013 at 10:49 AM, Andreas Bruckbauer <[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> So how does the manufacturer know the size of the pixel? I guess this will
> be calibrated at setup of the system with a target of known size for one
> objective and then stored somewhere in the parameters and hopfully
> regularly checked. Do they then just change it proportionally for the other
> objectives? Sometimes objective magnification can be a little bit different
> than what is written on the barrel, epecially when the lens has a
> correction collar, did anyone check this?
>
> We had the case that a journal insisted on us giving the magnification in
> the figure caption, so we gave the objective magnification and insisted on
> having the scale bar in addition to that.
>
> Andreas
>
>
>
>
>
>
>
> -----Original Message-----
> From: Guy Cox <[hidden email]>
> To: CONFOCALMICROSCOPY <[hidden email]>
> Sent: Fri, 16 Aug 2013 13:01
> Subject: Re: magnification
>
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> In a confocal image the magnification depends on the objective
> magnification and
> the area scanned on the sample.  There is of course an 'ocular' or
> transfer lens
> but that is not variable and thus is included in the calibration of the
> system.
> So when you 'zoom' in a confocal image you are just scanning a smaller
> part of
> the field of view.  The system should still always give you the correct
> magnification, but since this is a digital image the appropriate units are
> pixels per micrometre (or the converse, the size of a pixel).  Every
> confocal
> microscope will give you these figures.  You should check once in a while,
> of
> course, using a standard specimen such as a stage micrometer.
>
>                                                         Guy
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On
> Behalf Of Nicola Green
> Sent: Friday, 16 August 2013 8:57 PM
> To: [hidden email]
> Subject: magnification
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi
> This is probably a very foolish and basic question, but I am going to ask
> it any way and just apologise for my ignorance as a biological specialist
> using a confocal microscope.
>
> In microscopy I would generally multiply the objective lens magnification
> by the ocular lens (often 10x) to get the total magnification of an image.
> I want to know if this also applies for the confocal microscope or is the
> light path such that the objective magnification is the only one relevant?
> If I do need to include additional magnification what would these
> magnifications be, are they dependent upon the system being used (LSM 510
> META)  or is there a standard magnification?
>
> Thanks for your help in clarifying this for me.
>
> Regards
> Nicola
>
>
>
Chris Tully-2 Chris Tully-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hello All,

On wide field microscopes, the magnification (and calibration data ) can
be calculated with remarkable accuracy! If any of you are familiar with
Image-Pro Plus v5 or later, it includes a calibration estimation tool
that I helped develop and tested extensively against every microscope I
could get my hands on at the time.  Rarely were my calculated
calibrations off by more than +/-1%.

However, as has been pointed out repeatedly the magnification of an
image from a confocal is much more variable (I will skip the reasons
why). With that said magnification is simply:
                                        Image Size
Magnification = ---------------------------------
                                Original object size

So, with a scale bar that shows a known size the magnification can be
calculated at any time with a ruler and calculator.

In response to one posting about a journal editor insisting on listing
the magnification in the image discussion I think that we should all
push back pointing out that in today's world where it is easy, even
trivial to capture an image at any resolution and print it out in sizes
ranging from a few inches to many feet that a single number for
magnification is irrelevant and even misleading. To interpret a
statement of magnification I need to know many details of the way in
which the image presented was captured and manipulated post acquisition
(I do consider even cropping an image "manipulation").  But a scale bar
applied by the acquisition software or post acquisition based on a
spatial calibration gives me instant access to the scale of objects in
the image and if I so desire to the real magnification (i.e the
magnification after all manipulations that lead to it appearing on the
page in front of me). Further since many of now read articles in PDF
format, magnification becomes even squishier because I can used any PDF
viewer to zoom in on a single pixel should I so desire!

Chris Tully

On 8/16/2013 3:32 PM, Craig Brideau wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I agree.  With a confocal image the scale bar is really the only useful
> descriptor.  'Magnification' for confocals is more a function of the timing
> of the acquisition electronics, and talking about it in ocular terms is
> misleading.
>
> Craig
>
>
> On Fri, Aug 16, 2013 at 10:49 AM, Andreas Bruckbauer <[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> So how does the manufacturer know the size of the pixel? I guess this will
>> be calibrated at setup of the system with a target of known size for one
>> objective and then stored somewhere in the parameters and hopfully
>> regularly checked. Do they then just change it proportionally for the other
>> objectives? Sometimes objective magnification can be a little bit different
>> than what is written on the barrel, epecially when the lens has a
>> correction collar, did anyone check this?
>>
>> We had the case that a journal insisted on us giving the magnification in
>> the figure caption, so we gave the objective magnification and insisted on
>> having the scale bar in addition to that.
>>
>> Andreas
>>
>>
>>
>>
>>
>>
>>
>> -----Original Message-----
>> From: Guy Cox <[hidden email]>
>> To: CONFOCALMICROSCOPY <[hidden email]>
>> Sent: Fri, 16 Aug 2013 13:01
>> Subject: Re: magnification
>>
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> In a confocal image the magnification depends on the objective
>> magnification and
>> the area scanned on the sample.  There is of course an 'ocular' or
>> transfer lens
>> but that is not variable and thus is included in the calibration of the
>> system.
>> So when you 'zoom' in a confocal image you are just scanning a smaller
>> part of
>> the field of view.  The system should still always give you the correct
>> magnification, but since this is a digital image the appropriate units are
>> pixels per micrometre (or the converse, the size of a pixel).  Every
>> confocal
>> microscope will give you these figures.  You should check once in a while,
>> of
>> course, using a standard specimen such as a stage micrometer.
>>
>>                                                          Guy
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]]
>> On
>> Behalf Of Nicola Green
>> Sent: Friday, 16 August 2013 8:57 PM
>> To: [hidden email]
>> Subject: magnification
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi
>> This is probably a very foolish and basic question, but I am going to ask
>> it any way and just apologise for my ignorance as a biological specialist
>> using a confocal microscope.
>>
>> In microscopy I would generally multiply the objective lens magnification
>> by the ocular lens (often 10x) to get the total magnification of an image.
>> I want to know if this also applies for the confocal microscope or is the
>> light path such that the objective magnification is the only one relevant?
>> If I do need to include additional magnification what would these
>> magnifications be, are they dependent upon the system being used (LSM 510
>> META)  or is there a standard magnification?
>>
>> Thanks for your help in clarifying this for me.
>>
>> Regards
>> Nicola
>>
>>
>>


--
*Chris Tully*
Principal Consultant
240-475-9753


      Image Incyte, LLC

<http:%5C%5Cwww.ImageIncyte.com> [hidden email]
<mailto:[hidden email]>
Evelyn Ralston-2 Evelyn Ralston-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: magnification

In reply to this post by Nicola Green
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Nicola,

When you export your image from the 510 as a TIFF file, the software should give you the option to add a scale bar to the image. Ask a Zeiss rep if you don't find it. Alternately, your lab may own or borrow a graticule (also referred to as reticule or stage micrometer), i.e. a glass slide engraved with a 1 or 2mm scale, with marks for 10 and 100 microns intervals. Take a transmitted light image of the scale using each of the lenses present on the confocal and with a set zoom factor (important). You can then store these images and go back to the appropriate one to create a scale bar for any image, past or future. Just don't forget to take the zoom factor into account.

Evelyn Ralston, Ph.D.
Light Imaging Section
NIAMS, National Institutes of Health,
Bethesda, MD 20892





On Aug 16, 2013, at 6:56 AM, Nicola Green wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi
> This is probably a very foolish and basic question, but I am going to ask
> it any way and just apologise for my ignorance as a biological specialist
> using a confocal microscope.
>
> In microscopy I would generally multiply the objective lens magnification
> by the ocular lens (often 10x) to get the total magnification of an image.
> I want to know if this also applies for the confocal microscope or is the
> light path such that the objective magnification is the only one relevant?
> If I do need to include additional magnification what would these
> magnifications be, are they dependent upon the system being used (LSM 510
> META)  or is there a standard magnification?
>
> Thanks for your help in clarifying this for me.
>
> Regards
> Nicola