mounting medium for TIRF

classic Classic list List threaded Threaded
♦
♦
5 messages Options
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

I'd like to mount fixed cells on a coverslip to observe them by TIRF. Usually I use Mowiol to mount fixed cells but I realized that the refractive index gap necessary for TIRF between the glass and the medium is not present in Mowiol. What medium with a refraction index similar to water can I use ? Is PBS/Glycerol OK for TIRF ? On a related note, I hate nail polish, how can i seal PBS/gly mounted coverslips ?

Thank you for your advices

Christophe Leterrier
Kilgore, Jason Kilgore, Jason
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Sealing your coverslip can be accomplished with molten household wax, heated in a beaker or paraffin heater and applied with a Qtip applicator, or use molten VALAP.
 
 
Cheers,
 
Jason


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: Friday, February 08, 2008 8:30 AM
To: [hidden email]
Subject: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

I'd like to mount fixed cells on a coverslip to observe them by TIRF. Usually I use Mowiol to mount fixed cells but I realized that the refractive index gap necessary for TIRF between the glass and the medium is not present in Mowiol. What medium with a refraction index similar to water can I use ? Is PBS/Glycerol OK for TIRF ? On a related note, I hate nail polish, how can i seal PBS/gly mounted coverslips ?

Thank you for your advices

Christophe Leterrier
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mounting medium for TIRF

In reply to this post by lechristophe
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Zeiss has a nice, thought expensive, oil with water RI.  An alternative to nail polish is Permount, the stuff histological preparations are mounted with.  It can be diluted some with xylene if necessary, is clearly immiscible in water and forms a very tight moisture barrier.  It smells when fresh, though, so should be dried in a hood.
Carl
 
Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
Sent: Friday, February 08, 2008 9:30 AM
Subject: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

I'd like to mount fixed cells on a coverslip to observe them by TIRF. Usually I use Mowiol to mount fixed cells but I realized that the refractive index gap necessary for TIRF between the glass and the medium is not present in Mowiol. What medium with a refraction index similar to water can I use ? Is PBS/Glycerol OK for TIRF ? On a related note, I hate nail polish, how can i seal PBS/gly mounted coverslips ?

Thank you for your advices

Christophe Leterrier
Jennifer Waters Jennifer Waters
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
PBS & glycerol will have too high a refractive index for TIRF.  For fixed work we usually just use PBS plus an anti-fade like n-propyl galate (0.1%).  I also like VALAP (1:1:1 vaseline:lanolin:parafin) for sealing slides.  Use as a warmed liquid, solidifies at room temp.

Jennifer


On Feb 8, 2008 12:25 PM, Carl Boswell <[hidden email]> wrote:
Zeiss has a nice, thought expensive, oil with water RI.  An alternative to nail polish is Permount, the stuff histological preparations are mounted with.  It can be diluted some with xylene if necessary, is clearly immiscible in water and forms a very tight moisture barrier.  It smells when fresh, though, so should be dried in a hood.
Carl
 
Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
Sent: Friday, February 08, 2008 9:30 AM
Subject: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

I'd like to mount fixed cells on a coverslip to observe them by TIRF. Usually I use Mowiol to mount fixed cells but I realized that the refractive index gap necessary for TIRF between the glass and the medium is not present in Mowiol. What medium with a refraction index similar to water can I use ? Is PBS/Glycerol OK for TIRF ? On a related note, I hate nail polish, how can i seal PBS/gly mounted coverslips ?

Thank you for your advices

Christophe Leterrier



--
Jennifer Waters, Ph.D.
Director, Nikon Imaging Center at Harvard Medical School
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Thank you for your advices, I'll use the PBS moutning +VALAP sealing method to mount my coverlips. The question is : how do you keep your PBS+VALAP slides for say more than one week ?

- At 4°C it will allow growth of contaminants (bacteria, fungi...) relatively rapidly. Is it posssible to add something like sodium azide ? Related question : sodium azide is banned here in France, so do you know another efficient conservative ? I usually keep my fixed cells in 0.1% PFA for up to one or two weeks before immunostaining, but I think PFA will be autofluorescent if I put it in the mounting PBS as a conservative.

- If I freeze the slides, will the PBS make cristals and damage the cells ? How many freeze/thaw are possible for immunostained cultured cells without compromizong the cells/killing the fluorescence ?

Thanks for your help,

Christophe Leterrier




On Fri, Feb 8, 2008 at 20:06, Jennifer Waters <[hidden email]> wrote:
PBS & glycerol will have too high a refractive index for TIRF.  For fixed work we usually just use PBS plus an anti-fade like n-propyl galate (0.1%).  I also like VALAP (1:1:1 vaseline:lanolin:parafin) for sealing slides.  Use as a warmed liquid, solidifies at room temp.

Jennifer



On Feb 8, 2008 12:25 PM, Carl Boswell <[hidden email]> wrote:
Zeiss has a nice, thought expensive, oil with water RI.  An alternative to nail polish is Permount, the stuff histological preparations are mounted with.  It can be diluted some with xylene if necessary, is clearly immiscible in water and forms a very tight moisture barrier.  It smells when fresh, though, so should be dried in a hood.
Carl
 
Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
Sent: Friday, February 08, 2008 9:30 AM
Subject: mounting medium for TIRF

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear all,

I'd like to mount fixed cells on a coverslip to observe them by TIRF. Usually I use Mowiol to mount fixed cells but I realized that the refractive index gap necessary for TIRF between the glass and the medium is not present in Mowiol. What medium with a refraction index similar to water can I use ? Is PBS/Glycerol OK for TIRF ? On a related note, I hate nail polish, how can i seal PBS/gly mounted coverslips ?

Thank you for your advices

Christophe Leterrier



--
Jennifer Waters, Ph.D.
Director, Nikon Imaging Center at Harvard Medical School