To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen and carbon dioxide levels to the deep parts of the tissue we bubble the medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they are vascularized. Now we want to do the experiments with serum added to the medium, but the extra protein causes the bubbling medium to foam too much. Any suggestions on how to increase the dissolved O2 and CO2 without the foaming? Also, any ideas on how to monitor the resulting O2 levels? Thanks so much. Paul Herzmark [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Not talking from experience, but essentially bubbling is just to increase the surface in contact with the atmosphere for gas exchange and equilibrium to be reached. This should be possible to do by mearly having the media in the desired atmosphere with as much surface as possible exposed. This will probably not be as effective as bubbling but should work non the less. To speed up things I would design a (small box or scale up as necessary) container that is constantly supplied with the atmosphere and along the bottom a shallow but wide laminar flow of the medium (slightly tilting the box may be god for the flow). For monitoring the O2 levels an oxygen optode should do the trick. /Ricardo Figueroa On Tue, Jun 19, 2012 at 1:08 AM, Paul Herzmark <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen > and carbon dioxide levels to the deep parts of the tissue we bubble the > medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they > are vascularized. > > Now we want to do the experiments with serum added to the medium, but the > extra protein causes the bubbling medium to foam too much. > > Any suggestions on how to increase the dissolved O2 and CO2 without the > foaming? > Also, any ideas on how to monitor the resulting O2 levels? > > Thanks so much. > > > Paul Herzmark > > [hidden email] > Department of Molecular and Cell Biology > 479 Life Science Addition > University of California, Berkeley > Berkeley, CA 94720-3200 > (510) 643-9603 > |
![]() |
In reply to this post by Paul Herzmark
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Paul, when I was still in the lab we were monitoring oxygen concentration in media using an Oxylite Probe and Monitor like that provided by Oxford Optronix http://www.oxford-optronix.com/pO2monitors.htm(no commercial interest). We had this equipment because it is more commonly used for measuring intra-tumoral oxygen in patient but we found it effective for diffused Oxygen in media as well. The probes are expensive but flexible enough to thread through a small bore perfusion inlet. It worked very well but again, it requires an upfront investment in the equipment. Perhaps there are more cost-effective ways of doing this, we were lucky to have the equipment handy. Best Farid On Mon, Jun 18, 2012 at 4:08 PM, Paul Herzmark <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen > and carbon dioxide levels to the deep parts of the tissue we bubble the > medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they > are vascularized. > > Now we want to do the experiments with serum added to the medium, but the > extra protein causes the bubbling medium to foam too much. > > Any suggestions on how to increase the dissolved O2 and CO2 without the > foaming? > Also, any ideas on how to monitor the resulting O2 levels? > > Thanks so much. > > > Paul Herzmark > > [hidden email] > Department of Molecular and Cell Biology > 479 Life Science Addition > University of California, Berkeley > Berkeley, CA 94720-3200 > (510) 643-9603 > |
![]() |
In reply to this post by Paul Herzmark
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I don't know how your experimental set up works but have you tried using a stirrer bar and a plate to slowly stir the main media reservoir whilst gassing? You could probably avoid the foaming by turning down the gas flow. It should not affect the media as the stirring would be helping to distribute the gas evenly through the media. You may also want to fit proper air filters onto the media bottle which will help regulate the pressure inside the bottle which may assist. Regards, Deanne. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Herzmark Sent: Tuesday, 19 June 2012 9:09 AM To: [hidden email] Subject: oxygenate medium ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen and carbon dioxide levels to the deep parts of the tissue we bubble the medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they are vascularized. Now we want to do the experiments with serum added to the medium, but the extra protein causes the bubbling medium to foam too much. Any suggestions on how to increase the dissolved O2 and CO2 without the foaming? Also, any ideas on how to monitor the resulting O2 levels? Thanks so much. Paul Herzmark [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Thank you Deanne, I will try those ideas. I did get a dissolved O2 meter from Milwaukee Instruments. Unfortunately when I gas the medium with 95% O2 the level is higher than the meter reads. Now I need additional ideas! Paul On Thursday, July 12, 2012, Deanne Veronica Catmull wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I don't know how your experimental set up works but have you tried using a > stirrer bar and a plate to slowly stir the main media reservoir whilst > gassing? You could probably avoid the foaming by turning down the gas flow. > It should not affect the media as the stirring would be helping to > distribute the gas evenly through the media. You may also want to fit > proper air filters onto the media bottle which will help regulate the > pressure inside the bottle which may assist. > > Regards, > Deanne. > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]<javascript:;>] > On Behalf Of Paul Herzmark > Sent: Tuesday, 19 June 2012 9:09 AM > To: [hidden email] <javascript:;> > Subject: oxygenate medium > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen > and carbon dioxide levels to the deep parts of the tissue we bubble the > medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they > are vascularized. > > Now we want to do the experiments with serum added to the medium, but the > extra protein causes the bubbling medium to foam too much. > > Any suggestions on how to increase the dissolved O2 and CO2 without the > foaming? > Also, any ideas on how to monitor the resulting O2 levels? > > Thanks so much. > > > Paul Herzmark > > [hidden email] <javascript:;> > Department of Molecular and Cell Biology > 479 Life Science Addition > University of California, Berkeley > Berkeley, CA 94720-3200 > (510) 643-9603 > -- Paul Herzmark Specialist [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 (510) 643-9500 fax |
![]() |
In reply to this post by Deanne Veronica Catmull
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Paul, this is just an idea, haven't tried it- make an artificial gill: use a blood dialysis cartridge and make two independent fluid circuits: one closed loop in which you insert a gas-wash bottle (so it will foam in this circuit), and one that flows through the fibers into your observation chamber. Just make sure that both liquids are identical and fresh. Best regards, Thomas Dr. Thomas Horn, The Single Cell Unit, U1.46 Department of Biosystems Science and Engineering (D-BSSE) Swiss Federal Institute of Technology Zurich (ETH) Mattenstrasse 26 CH 4048 Basel Switzerland Phone: +41 61 387 3373 mail: [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Deanne Veronica Catmull Sent: Freitag, 13. Juli 2012 07:46 To: [hidden email] Subject: Re: oxygenate medium ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I don't know how your experimental set up works but have you tried using a stirrer bar and a plate to slowly stir the main media reservoir whilst gassing? You could probably avoid the foaming by turning down the gas flow. It should not affect the media as the stirring would be helping to distribute the gas evenly through the media. You may also want to fit proper air filters onto the media bottle which will help regulate the pressure inside the bottle which may assist. Regards, Deanne. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Paul Herzmark Sent: Tuesday, 19 June 2012 9:09 AM To: [hidden email] Subject: oxygenate medium ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, We do 2P microscopy of thick tissue explants. To deliver appropriate oxygen and carbon dioxide levels to the deep parts of the tissue we bubble the medium with 95% O2 and 5% CO2. That keeps the cells as happy as when they are vascularized. Now we want to do the experiments with serum added to the medium, but the extra protein causes the bubbling medium to foam too much. Any suggestions on how to increase the dissolved O2 and CO2 without the foaming? Also, any ideas on how to monitor the resulting O2 levels? Thanks so much. Paul Herzmark [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Thomas, Thank you for that idea. I will look into it and report back! Paul Paul Herzmark Specialist [hidden email] Department of Molecular and Cell Biology 479 Life Science Addition University of California, Berkeley Berkeley, CA 94720-3200 (510) 643-9603 (510) 643-9500 fax On Tue, Jul 17, 2012 at 12:32 AM, Horn Thomas <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello Paul, > this is just an idea, haven't tried it- make an artificial gill: use a > blood dialysis cartridge and make two independent fluid circuits: one > closed loop in which you insert a gas-wash bottle (so it will foam in this > circuit), and one that flows through the fibers into your observation > chamber. > Just make sure that both liquids are identical and fresh. > Best regards, > Thomas > > > Dr. Thomas Horn, > The Single Cell Unit, U1.46 > Department of Biosystems Science and Engineering (D-BSSE) > Swiss Federal Institute of Technology Zurich (ETH) > Mattenstrasse 26 > CH 4048 Basel > Switzerland > Phone: +41 61 387 3373 > mail: [hidden email] > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Deanne Veronica Catmull > Sent: Freitag, 13. Juli 2012 07:46 > To: [hidden email] > Subject: Re: oxygenate medium > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I don't know how your experimental set up works but have you tried using a > stirrer bar and a plate to slowly stir the main media reservoir whilst > gassing? You could probably avoid the foaming by turning down the gas flow. > It should not affect the media as the stirring would be helping to > distribute the gas evenly through the media. You may also want to fit > proper air filters onto the media bottle which will help regulate the > pressure inside the bottle which may assist. > > Regards, > Deanne. > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Paul Herzmark > Sent: Tuesday, 19 June 2012 9:09 AM > To: [hidden email] > Subject: oxygenate medium > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > We do 2P microscopy of thick tissue explants. To deliver appropriate > oxygen and carbon dioxide levels to the deep parts of the tissue we bubble > the medium with 95% O2 and 5% CO2. That keeps the cells as happy as when > they are vascularized. > > Now we want to do the experiments with serum added to the medium, but the > extra protein causes the bubbling medium to foam too much. > > Any suggestions on how to increase the dissolved O2 and CO2 without the > foaming? > Also, any ideas on how to monitor the resulting O2 levels? > > Thanks so much. > > > Paul Herzmark > > [hidden email] > Department of Molecular and Cell Biology > 479 Life Science Addition > University of California, Berkeley > Berkeley, CA 94720-3200 > (510) 643-9603 > |