penetration depth

classic Classic list List threaded Threaded
♦
♦
17 messages Options
Sarah Kefayati Sarah Kefayati
Reply | Threaded
Open this post in threaded view
♦
♦
|

penetration depth

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello all!
 
I was wondering for which one penetration depth is higher: NA:1.2 60x lens or NA : 0.3 10x lens?
 
thanks
Sarah
Peng Xi-2 Peng Xi-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Sarah,
    The penetration depth is generally related with the working
distance, and of course, the NA 0.3 10X has much longer penetration
depth than that of NA1.2 60x. In general, NA1.2 60x has a working
distance of ~100 um, and NA0.3 10x has longer than 1 mm.
    But the penetration depth is also related with the scattering
coefficient of the sample. If you are observing semi-transparent
sample, then it is almost equal to working distance if omitted the
aberration; but if you are observing highly scattering sample, then it
can be the same because most light are simply scattered along the way.
    Thank you!
--
Best regards,
Peng Xi
Associate Professor
Institute for Laser Medicine and Biophotonics
Shanghai Jiao Tong University
800 Dongchuan Rd.
Shanghai 200240, China
Tel: (86) 21-3420-4076
http://biophotonics.sjtu.edu.cn/


On Feb 5, 2008 9:31 AM, Sarah Kefayati <[hidden email]> wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Hello all!
>
> I was wondering for which one penetration depth is higher: NA:1.2 60x lens
> or NA : 0.3 10x lens?
>
> thanks
> Sarah
Sarah Kefayati Sarah Kefayati
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Thank you so much for your great info,totally helpful!
 
Sarah

On Feb 4, 2008 8:53 PM, Peng Xi <[hidden email]> wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Sarah,
   The penetration depth is generally related with the working
distance, and of course, the NA 0.3 10X has much longer penetration
depth than that of NA1.2 60x. In general, NA1.2 60x has a working
distance of ~100 um, and NA0.3 10x has longer than 1 mm.
   But the penetration depth is also related with the scattering
coefficient of the sample. If you are observing semi-transparent
sample, then it is almost equal to working distance if omitted the
aberration; but if you are observing highly scattering sample, then it
can be the same because most light are simply scattered along the way.
   Thank you!
--
Best regards,
Peng Xi
Associate Professor
Institute for Laser Medicine and Biophotonics
Shanghai Jiao Tong University
800 Dongchuan Rd.
Shanghai 200240, China
Tel: (86) 21-3420-4076
http://biophotonics.sjtu.edu.cn/


On Feb 5, 2008 9:31 AM, Sarah Kefayati <[hidden email]> wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Hello all!
>
> I was wondering for which one penetration depth is higher: NA:1.2 60x lens
> or NA : 0.3 10x lens?
>
> thanks
> Sarah

Stephen Cody Stephen Cody
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear Sarah,

 

For best results, you actually should use a long working distance dipping lens for thick live (or hydrated specimens).

 

Cheers

Steve

Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research

PO Box 2008 Royal Melbourne Hospital
Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sarah Kefayati
Sent: Tuesday, 5 February 2008 12:59 PM
To: [hidden email]
Subject: Re: penetration depth

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Thank you so much for your great info,totally helpful!

 

Sarah

On Feb 4, 2008 8:53 PM, Peng Xi <[hidden email]> wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Sarah,
   The penetration depth is generally related with the working
distance, and of course, the NA 0.3 10X has much longer penetration
depth than that of NA1.2 60x. In general, NA1.2 60x has a working
distance of ~100 um, and NA0.3 10x has longer than 1 mm.
   But the penetration depth is also related with the scattering
coefficient of the sample. If you are observing semi-transparent
sample, then it is almost equal to working distance if omitted the
aberration; but if you are observing highly scattering sample, then it
can be the same because most light are simply scattered along the way.
   Thank you!
--
Best regards,
Peng Xi
Associate Professor
Institute for Laser Medicine and Biophotonics
Shanghai Jiao Tong University
800 Dongchuan Rd.
Shanghai 200240, China
Tel: (86) 21-3420-4076
http://biophotonics.sjtu.edu.cn/


On Feb 5, 2008 9:31 AM, Sarah Kefayati <[hidden email]> wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

> Hello all!
>
> I was wondering for which one penetration depth is higher: NA:1.2 60x lens
> or NA : 0.3 10x lens?
>
> thanks
> Sarah

 


This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research.

Bill Miller-3 Bill Miller-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by Sarah Kefayati
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

ultimately the lower NA with its longer working distance wins if the
sample is clear enough..


At 08:31 PM 2/4/2008 -0500, you wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Hello all!
>
>I was wondering for which one penetration depth is higher: NA:1.2
>60x lens or NA : 0.3 10x lens?
>
>thanks
>Sarah
Judy Trogadis Judy Trogadis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

What's the best method of making a sample more transparent to reduce scattering, without changing its chemistry too much? (other than methyl salicylate and glycerol).

Thanks.
Judy

Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]


>>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

ultimately the lower NA with its longer working distance wins if the
sample is clear enough..


At 08:31 PM 2/4/2008 -0500, you wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
>Hello all!
>
>I was wondering for which one penetration depth is higher: NA:1.2
>60x lens or NA : 0.3 10x lens?
>
>thanks
>Sarah
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

TDE mounting medium

In reply to this post by Stephen Cody
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Does anyone in either Sydney or Perth have any of the
high-index water-miscible mounting medium TDE
(2,2`-thiodiethanol)?  We need some urgently for
a STED session later this month but unfortunately
Sigma Australia have no stock and because it's
toxic shipping from Germany will take several
weeks (which we don't have).

                                      Guy

 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net <http://www.guycox.net/>  


No virus found in this outgoing message.
Checked by AVG Free Edition.
Version: 7.5.516 / Virus Database: 269.19.19/1258 - Release Date: 4/02/2008 10:10 AM
 
Jeffrey L. Travis Jeffrey L. Travis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by Bill Miller-3
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Just a heads-up:  Emails with titles like "Penetration Depth"  generally
end up in my trash along with other "male enhancement" solicitations.  I
am just saying!

Jeff Travis

Bill Miller wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> ultimately the lower NA with its longer working distance wins if the
> sample is clear enough..
>
>
> At 08:31 PM 2/4/2008 -0500, you wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>> Hello all!
>>
>> I was wondering for which one penetration depth is higher: NA:1.2 60x
>> lens or NA : 0.3 10x lens?
>>
>> thanks
>> Sarah
>
Mayandi Sivaguru Mayandi Sivaguru
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Judy, try the BABB after Dent's fixative. BABB or Murray's clear stands
for 1:2 mixture of benzyl alcohol and benzyl benzoate both RI around 1.5
and available from Sigma. Let me offline should you have problem or
success with this protocol.
Thanks
Shiv

Mayandi Sivaguru PhD, PhD
Microscopy Facility Manager
Institute for Genomic Biology
Unviersity of Illinois at Urbaba-Champaign
Urbana 61801 IL USA
[hidden email]
http://core.igb.uiuc.edu


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> What's the best method of making a sample more transparent to reduce
> scattering, without changing its chemistry too much? (other than methyl
> salicylate and glycerol).
>
> Thanks.
> Judy
>
> Judy Trogadis
> Bio-Imaging Coordinator
> St. Michael's Hospital, 7Queen
> 30 Bond St.
> Toronto, ON M5B 1W8, Canada
> ph:  416-864-6060  x6337
> pager: 416-685-9219
> fax: 416-864-6043
> [hidden email]
>
>
>>>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> ultimately the lower NA with its longer working distance wins if the
> sample is clear enough..
>
>
> At 08:31 PM 2/4/2008 -0500, you wrote:
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>Hello all!
>>
>>I was wondering for which one penetration depth is higher: NA:1.2
>>60x lens or NA : 0.3 10x lens?
>>
>>thanks
>>Sarah
>
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

But, it will depend upon the highest RI component of the sample.  you  
might have to go higher or lower by blending.  I've used Arnie Voie's  
mixture of 5 parts methyl salicylate:3 parts benzyl benzoate to  
obtain an RI that calculates to about 1.556.   It works well for  
dense connective tissue of bone.

Regards,

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************
******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************
******


On Feb 6, 2008, at 7:11 AM, Mayandi Sivaguru wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Judy, try the BABB after Dent's fixative. BABB or Murray's clear  
> stands
> for 1:2 mixture of benzyl alcohol and benzyl benzoate both RI  
> around 1.5
> and available from Sigma. Let me offline should you have problem or
> success with this protocol.
> Thanks
> Shiv
>
> Mayandi Sivaguru PhD, PhD
> Microscopy Facility Manager
> Institute for Genomic Biology
> Unviersity of Illinois at Urbaba-Champaign
> Urbana 61801 IL USA
> [hidden email]
> http://core.igb.uiuc.edu
>
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> What's the best method of making a sample more transparent to reduce
>> scattering, without changing its chemistry too much? (other than  
>> methyl
>> salicylate and glycerol).
>>
>> Thanks.
>> Judy
>>
>> Judy Trogadis
>> Bio-Imaging Coordinator
>> St. Michael's Hospital, 7Queen
>> 30 Bond St.
>> Toronto, ON M5B 1W8, Canada
>> ph:  416-864-6060  x6337
>> pager: 416-685-9219
>> fax: 416-864-6043
>> [hidden email]
>>
>>
>>>>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> ultimately the lower NA with its longer working distance wins if the
>> sample is clear enough..
>>
>>
>> At 08:31 PM 2/4/2008 -0500, you wrote:
>>> Search the CONFOCAL archive at
>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>> Hello all!
>>>
>>> I was wondering for which one penetration depth is higher: NA:1.2
>>> 60x lens or NA : 0.3 10x lens?
>>>
>>> thanks
>>> Sarah
>>
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by Judy Trogadis
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Judy,

I was impressed with 2,2'-thiodiethanol, used neat (R.I. 1.52;
Sigma-Aldrich). Completely black background (N=2 slides, one confocal
session). See Staudt, Hell et al 2007 Microsc Res Tech for original
reference. They used RI 1.515 by adding 3% H2O. I figured no reason
to pollute the TDE with any additives. One oddity was that by eye,
both blood vessel elastin fiber autofluorescence was decreased (488
nm excitation, green emission) and the DAPI nuclear counterstain was
quenched. Both came through nicely in the confocal.

Robert Zucker has several confocal papers on BABB - benzyl
alcohol/benzyl benzoate as a clearing agent (see another responder's
message for other names). Bob told me the key is to completely
dehydrate the specimen through EtOH steps into xylene, before going
to BABB. See his papers for exact steps.

best wishes,

George


At 11:09 AM 2/5/2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>What's the best method of making a sample more transparent to reduce
>scattering, without changing its chemistry too much? (other than
>methyl salicylate and glycerol).
>
>Thanks.
>Judy
>
>Judy Trogadis
>Bio-Imaging Coordinator
>St. Michael's Hospital, 7Queen
>30 Bond St.
>Toronto, ON M5B 1W8, Canada
>ph:  416-864-6060  x6337
>pager: 416-685-9219
>fax: 416-864-6043
>[hidden email]
>
>
> >>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>ultimately the lower NA with its longer working distance wins if the
>sample is clear enough..
>
>
>At 08:31 PM 2/4/2008 -0500, you wrote:
> >Search the CONFOCAL archive at
> >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> >Hello all!
> >
> >I was wondering for which one penetration depth is higher: NA:1.2
> >60x lens or NA : 0.3 10x lens?
> >
> >thanks
> >Sarah






George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc 
(Analytical Imaging Core Facility)
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear George,
Did Bob say why the xylene was important?  I would  expect that BABB  
would be as alcohol-soluble as methyl salicylate:benzyl benzoate.  
Maybe the xylene was ensuring the last of the water is out, but  
thorough dehydration should be sufficient.  Many mixtures of solvents  
are given in this book: Spalteholz, W., 1914. Über das  
Durchsichtigmachen von menschlichen und tierischen Präparaten und  
seine theoretischen Bedingungen. Hirzel, Leipzig.

Glad to hear someone else has tried TDE. Do you notice any tissue  
shrinkage? How are you storing it?  I've haven't yet opened my  
bottle.  One post doc is using my mixture of MSBB on 1200 µm brain  
slices, but I'm going to try TDE on brain slices <400 µm with grad  
student. Thinner slices tend to curl after dehydration.

Regards,
Glen

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************
******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************
******


On Feb 7, 2008, at 6:58 PM, George McNamara wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Judy,
>
> I was impressed with 2,2'-thiodiethanol, used neat (R.I. 1.52;  
> Sigma-Aldrich). Completely black background (N=2 slides, one  
> confocal session). See Staudt, Hell et al 2007 Microsc Res Tech for  
> original reference. They used RI 1.515 by adding 3% H2O. I figured  
> no reason to pollute the TDE with any additives. One oddity was  
> that by eye, both blood vessel elastin fiber autofluorescence was  
> decreased (488 nm excitation, green emission) and the DAPI nuclear  
> counterstain was quenched. Both came through nicely in the confocal.
>
> Robert Zucker has several confocal papers on BABB - benzyl alcohol/
> benzyl benzoate as a clearing agent (see another responder's  
> message for other names). Bob told me the key is to completely  
> dehydrate the specimen through EtOH steps into xylene, before going  
> to BABB. See his papers for exact steps.
>
> best wishes,
>
> George
>
>
> At 11:09 AM 2/5/2008, you wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> What's the best method of making a sample more transparent to  
>> reduce scattering, without changing its chemistry too much? (other  
>> than methyl salicylate and glycerol).
>>
>> Thanks.
>> Judy
>>
>> Judy Trogadis
>> Bio-Imaging Coordinator
>> St. Michael's Hospital, 7Queen
>> 30 Bond St.
>> Toronto, ON M5B 1W8, Canada
>> ph:  416-864-6060  x6337
>> pager: 416-685-9219
>> fax: 416-864-6043
>> [hidden email]
>>
>>
>> >>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> ultimately the lower NA with its longer working distance wins if the
>> sample is clear enough..
>>
>>
>> At 08:31 PM 2/4/2008 -0500, you wrote:
>> >Search the CONFOCAL archive at
>> >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>> >Hello all!
>> >
>> >I was wondering for which one penetration depth is higher: NA:1.2
>> >60x lens or NA : 0.3 10x lens?
>> >
>> >thanks
>> >Sarah
>
>
>
>
>
>
> George McNamara, Ph.D.
> University of Miami, Miller School of Medicine
> Image Core
> Miami, FL 33010
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://home.earthlink.net/~pubspectra/
> http://home.earthlink.net/~geomcnamara/
> http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc 
> (Analytical Imaging Core Facility)
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Glen,

Bob Zucker told me that if the tissue still has
any water in it, the BABB will turn cloudy.
Another item to be aware of with BABB is that it
can solubilize some glues, so using it in an
imaging chamber could results in leaks. Bob also
emphasized that BABB is a very bad thing to have
land on an objective lens or any other optic. One
last item about BABB - I found some chemical
company web sites that listed the refractive
indices of several benzyl X compounds. Some are
at or below RI=1.515. So, it may be feasible to
find a mixture of two that is right at 1.515.
That said, I'm excited about TDE because even
neat, its RI is close to that of immersion oil
and glass as to not matter except maybe hyper-critical deconvolution.

I am currently storing the TDE (25 mL bottle,
smallest that Sigma-Aldrich sells) in the
chemical cabinet. I will probably aliquot it out
into 0.5 or 1.0 mL eppendorf tubes to make it
easier to distribute to users who want to try.

My first user had ~5 um lung tissue sections. No
way for me to tell whether they had shrinkage.
There was an excellent article in one of the
trade magazines (Microscopy Today?) a year or so
on paraformaldehyde concentration in perfusion
fixation. That, and other articles, lead me to
expect that shrinkage comes from the PFA step,
not from the later tissue clearing and dehydration.

George



At 11:25 AM 2/8/2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Dear George,
>Did Bob say why the xylene was important?  I would  expect that BABB
>would be as alcohol-soluble as methyl salicylate:benzyl benzoate.
>Maybe the xylene was ensuring the last of the water is out, but
>thorough dehydration should be sufficient.  Many mixtures of solvents
>are given in this book: Spalteholz, W., 1914. Über das
>Durchsichtigmachen von menschlichen und tierischen Präparaten und
>seine theoretischen Bedingungen. Hirzel, Leipzig.
>
>Glad to hear someone else has tried TDE. Do you notice any tissue
>shrinkage? How are you storing it?  I've haven't yet opened my
>bottle.  One post doc is using my mixture of MSBB on 1200 µm brain
>slices, but I'm going to try TDE on brain slices <400 µm with grad
>student. Thinner slices tend to curl after dehydration.
>
>Regards,
>Glen
>
>Glen MacDonald
>Core for Communication Research
>Virginia Merrill Bloedel Hearing Research Center
>Box 357923
>University of Washington
>Seattle, WA 98195-7923  USA
>(206) 616-4156
>[hidden email]
>
>************************************************************************
>******
>The box said "Requires WindowsXP or better", so I bought a Macintosh.
>************************************************************************
>******
>
>
>On Feb 7, 2008, at 6:58 PM, George McNamara wrote:
>
>>Search the CONFOCAL archive at
>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>Hi Judy,
>>
>>I was impressed with 2,2'-thiodiethanol, used neat (R.I. 1.52;
>>Sigma-Aldrich). Completely black background (N=2 slides, one
>>confocal session). See Staudt, Hell et al 2007 Microsc Res Tech for
>>original reference. They used RI 1.515 by adding 3% H2O. I figured
>>no reason to pollute the TDE with any additives. One oddity was
>>that by eye, both blood vessel elastin fiber autofluorescence was
>>decreased (488 nm excitation, green emission) and the DAPI nuclear
>>counterstain was quenched. Both came through nicely in the confocal.
>>
>>Robert Zucker has several confocal papers on
>>BABB - benzyl alcohol/ benzyl benzoate as a
>>clearing agent (see another responder's
>>message for other names). Bob told me the key is to completely
>>dehydrate the specimen through EtOH steps into xylene, before going
>>to BABB. See his papers for exact steps.
>>
>>best wishes,
>>
>>George
>>
>>
>>At 11:09 AM 2/5/2008, you wrote:
>>>Search the CONFOCAL archive at
>>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>>What's the best method of making a sample more transparent to
>>>reduce scattering, without changing its chemistry too much? (other
>>>than methyl salicylate and glycerol).
>>>
>>>Thanks.
>>>Judy
>>>
>>>Judy Trogadis
>>>Bio-Imaging Coordinator
>>>St. Michael's Hospital, 7Queen
>>>30 Bond St.
>>>Toronto, ON M5B 1W8, Canada
>>>ph:  416-864-6060  x6337
>>>pager: 416-685-9219
>>>fax: 416-864-6043
>>>[hidden email]
>>>
>>>
>>> >>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
>>>Search the CONFOCAL archive at
>>>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>>ultimately the lower NA with its longer working distance wins if the
>>>sample is clear enough..
>>>
>>>
>>>At 08:31 PM 2/4/2008 -0500, you wrote:
>>> >Search the CONFOCAL archive at
>>> >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>> >Hello all!
>>> >
>>> >I was wondering for which one penetration depth is higher: NA:1.2
>>> >60x lens or NA : 0.3 10x lens?
>>> >
>>> >thanks
>>> >Sarah
>>
>>
>>
>>
>>
>>
>>George McNamara, Ph.D.
>>University of Miami, Miller School of Medicine
>>Image Core
>>Miami, FL 33010
>>[hidden email]
>>[hidden email]
>>305-243-8436 office
>>http://home.earthlink.net/~pubspectra/
>>http://home.earthlink.net/~geomcnamara/
>>http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc
>>(Analytical Imaging Core Facility)






George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc 
(Analytical Imaging Core Facility)
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by Sarah Kefayati
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: penetration depth
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Glen,

Bob Zucker told me that if the tissue still has any water in it, the BABB will turn cloudy. Another item to be aware of with BABB is that it can solubilize some glues, so using it in an imaging chamber could results in leaks. Bob also emphasized that BABB is a very bad thing to have land on an objective lens or any other optic. One last item about BABB - I found some chemical company web sites that listed the refractive indices of several benzyl X compounds. Some are at or below RI=1.515. So, it may be feasible to find a mixture of two that is right at 1.515. That said, I'm excited about TDE because even neat, its RI is close to that of immersion oil and glass as to not matter except maybe hyper-critical deconvolution.

I am currently storing the TDE (25 mL bottle, smallest that Sigma-Aldrich sells) in the chemical cabinet. I will probably aliquot it out into 0.5 or 1.0 mL eppendorf tubes to make it easier to distribute to users who want to try.

My first user had ~5 um lung tissue sections. No way for me to tell whether they had shrinkage. There was an excellent article in one of the trade magazines (Microscopy Today?) a year or so on paraformaldehyde concentration in perfusion fixation. That, and other articles, lead me to expect that shrinkage comes from the PFA step, not from the later tissue clearing and dehydration.
George


Hi George,

On your last point: shrinkage

In the late 1970s Alan Boyde and Elaine Maconnachie published a number of papers (some noted below) describing experiments where they followed the size of various specimens (fetal rat limb buds etc.) during all the transitions between being alive and being an SEM specimen (i.e., fixation, dehydration, replacement with intermediate and transition liquids, actual drying). These images where then analyzed quantitatively "on-line"  with a Quantimet image analyzing computer.

They used a wide variety of CDP protocols (different dehydrations, different intermediate and transition liquids) as well as freeze-drying from water ice and also from "ices" made by freezing specimens immersed in non-polar liquids.

Although every type of transition could produced changes in volume, by far the largest changes were always associated with the final steps of dehydration (say from 80% to 100% ethanol or acetone).

I should also point out that the total shrinkage was "massive".  Although freeze drying was best (about 40% VOLUME shrinkage as I recall, most if it occurring  as the "last of the last" of the ice or bound-water was removed as the specimen warmed up in the vacuum) , the best CPD protocols caused "only" about 60% VOLUME shrinkage. Others produced 80% shrinkage.

How could this be one might ask. Surely SEM specimens could not look so "real" had they shrunk to this degree. Part of the answer has to do with the fact that many SEM (and LM) specimens consist of cells adhered to glass or plastic substrates. The glass does not shrink. The cells attached to it become thinner, rather like the skin of an animal pegged out on a frame and left in the sun to dry. As we seldom take the trouble to measure their thickness, its reduction goes unremarked.

In Bob Bacallao's Chapter 18 in the Handbook, one can see images of this effect. On the other hand, the limb buds were merely suspended in liquid, not constrained in any way.  In addition, dehydration protocols often extract lipids and there my be more of these in limb buds than in the average tissue culture cell. Finally, cells with high water content shrunk more.

The basic message is the same: The proteins shrivel up as the last of the water is removed, whether this occurs by by sublimation or by dehydration.

I very much suspect that the same will be true of specimens embedded in clearing agents.

Cheers,

Jim P.

Boyde, A., and Maconnachie, E., 1979, "Volume changes during preparation
of mouse embryonic tissue for scanning electron microscopy." Scanning
2:149-163.
Boyde, A., and Maconnachie, E., 1981, "Morphological correlations with
dimensional change during SEM specimen preparation." Scanning Electron
Microsc. 4:278-34.
-- 
              **********************************************
Prof. James B. Pawley,                                          Ph.  608-263-3147 
Room 223, Zoology Research Building,                                  FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706                                [hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/            Applications due by March 15, 2008
       "If it ain't diffraction, it must be statistics." Anon.
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth ... tissue shrinkage

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Jim,

The article I referred to is:

C.W. Scouten, R. O'Connor, M. Cunningham 2006 Perfusion fixation of research animal. Microscopy Today May 2006: 26-33. MT low resolution issues are posted online somewhere.

Two quotes:

"Intracellular fluid of living cells usually fluctuates about 330 millMolar (mM) osmolality. By contrast, four percent formaldehyde in water, commonly used in fixation, is about 1400 mM osmolality."

"Several lines of evidence ... show that living brain is about 20% extracellular space. In perfused and fixed tissue, using traditional protocols, this space is absent and the brain is reduced in volume by about 20%. ... The brain atlas by Paxinos and Watson (1998) avoided this problem by working only with fresh frozen tissue, and not fixing."

The paper goes on to define better perfusion reagents and protocols. I wish it had been published in a peer reviewed journal instead of a "gray literature" trade magazine, and hope the authors publish in a real journal.

Sample preparation artifacts, such as the publications you mention, are among the reasons why I encourage users to image their specimens immediate after sacrificing the animal, with careful attention to perfusion fixation issues such as those raised by Scouten et al. Hoechst dyes only take minutes to soak into even an intact mouse brain (multiphoton). DiI, DiO, etc, can be used to label all the endothelial cells (both sides, including angiogenic sprouts) prior to perfusion fixation (Rong Wen, U Miami, in prep). After the tissue has been imaged by confocal and/or multiphoton microscopy (optionally with slicing into confocal penetration depth compatible thickness), it can be brought to the lab and processed for sectioning, immunofluorescence, in situ hybridization, H&E, IHC, etc.

George




At 12:30 PM 2/10/2008, you wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Glen,

Bob Zucker told me that if the tissue still has any water in it, the BABB will turn cloudy. Another item to be aware of with BABB is that it can solubilize some glues, so using it in an imaging chamber could results in leaks. Bob also emphasized that BABB is a very bad thing to have land on an objective lens or any other optic. One last item about BABB - I found some chemical company web sites that listed the refractive indices of several benzyl X compounds. Some are at or below RI=1.515. So, it may be feasible to find a mixture of two that is right at 1.515. That said, I'm excited about TDE because even neat, its RI is close to that of immersion oil and glass as to not matter except maybe hyper-critical deconvolution.

I am currently storing the TDE (25 mL bottle, smallest that Sigma-Aldrich sells) in the chemical cabinet. I will probably aliquot it out into 0.5 or 1.0 mL eppendorf tubes to make it easier to distribute to users who want to try.

My first user had ~5 um lung tissue sections. No way for me to tell whether they had shrinkage. There was an excellent article in one of the trade magazines (Microscopy Today?) a year or so on paraformaldehyde concentration in perfusion fixation. That, and other articles, lead me to expect that shrinkage comes from the PFA step, not from the later tissue clearing and dehydration.
George


Hi George,

On your last point: shrinkage

In the late 1970s Alan Boyde and Elaine Maconnachie published a number of papers (some noted below) describing experiments where they followed the size of various specimens (fetal rat limb buds etc.) during all the transitions between being alive and being an SEM specimen (i.e., fixation, dehydration, replacement with intermediate and transition liquids, actual drying). These images where then analyzed quantitatively "on-line"  with a Quantimet image analyzing computer.

They used a wide variety of CDP protocols (different dehydrations, different intermediate and transition liquids) as well as freeze-drying from water ice and also from "ices" made by freezing specimens immersed in non-polar liquids.

Although every type of transition could produced changes in volume, by far the largest changes were always associated with the final steps of dehydration (say from 80% to 100% ethanol or acetone).

I should also point out that the total shrinkage was "massive".  Although freeze drying was best (about 40% VOLUME shrinkage as I recall, most if it occurring  as the "last of the last" of the ice or bound-water was removed as the specimen warmed up in the vacuum) , the best CPD protocols caused "only" about 60% VOLUME shrinkage. Others produced 80% shrinkage.

How could this be one might ask. Surely SEM specimens could not look so "real" had they shrunk to this degree. Part of the answer has to do with the fact that many SEM (and LM) specimens consist of cells adhered to glass or plastic substrates. The glass does not shrink. The cells attached to it become thinner, rather like the skin of an animal pegged out on a frame and left in the sun to dry. As we seldom take the trouble to measure their thickness, its reduction goes unremarked.

In Bob Bacallao's Chapter 18 in the Handbook, one can see images of this effect. On the other hand, the limb buds were merely suspended in liquid, not constrained in any way.  In addition, dehydration protocols often extract lipids and there my be more of these in limb buds than in the average tissue culture cell. Finally, cells with high water content shrunk more.

The basic message is the same: The proteins shrivel up as the last of the water is removed, whether this occurs by by sublimation or by dehydration.

I very much suspect that the same will be true of specimens embedded in clearing agents.

Cheers,

Jim P.

Boyde, A., and Maconnachie, E., 1979, "Volume changes during preparation
of mouse embryonic tissue for scanning electron microscopy." Scanning
2:149-163.
Boyde, A., and Maconnachie, E., 1981, "Morphological correlations with
dimensional change during SEM specimen preparation." Scanning Electron
Microsc.
4:278-34.


-- 
              **********************************************
Prof. James B. Pawley,                                        Ph.  608-263-3147
Room 223, Zoology Research Building,                                 FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706                              [hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/              Applications due by March 15, 2008
                "If it ain't diffraction, it must be statistics." Anon.




 

George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc (Analytical Imaging Core Facility)


Ron Anderson-4 Ron Anderson-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth ... tissue shrinkage

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

You can download the article at http://www.microscopy-today.com click on
tables of contents/2006/ and then the May issue.

Ron Anderson, MT Editor
Subscribe to MT at the same website, free in North America. Only my hair
is gray.

George McNamara wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Jim,
>
> The article I referred to is:
>
> C.W. Scouten, R. O'Connor, M. Cunningham 2006 Perfusion fixation of
> research animal. Microscopy Today May 2006: 26-33. MT low resolution
> issues are posted online somewhere.
>
> Two quotes:
>
> "Intracellular fluid of living cells usually fluctuates about 330
> millMolar (mM) osmolality. By contrast, four percent formaldehyde in
> water, commonly used in fixation, is about 1400 mM osmolality."
>
> "Several lines of evidence ... show that living brain is about 20%
> extracellular space. In perfused and fixed tissue, using traditional
> protocols, this space is absent and the brain is reduced in volume by
> about 20%. ... The brain atlas by Paxinos and Watson (1998) avoided
> this problem by working only with fresh frozen tissue, and not fixing."
>
> The paper goes on to define better perfusion reagents and protocols. I
> wish it had been published in a peer reviewed journal instead of a
> "gray literature" trade magazine, and hope the authors publish in a
> real journal.
>
> Sample preparation artifacts, such as the publications you mention,
> are among the reasons why I encourage users to image their specimens
> immediate after sacrificing the animal, with careful attention to
> perfusion fixation issues such as those raised by Scouten et al.
> Hoechst dyes only take minutes to soak into even an intact mouse brain
> (multiphoton). DiI, DiO, etc, can be used to label all the endothelial
> cells (both sides, including angiogenic sprouts) prior to perfusion
> fixation (Rong Wen, U Miami, in prep). After the tissue has been
> imaged by confocal and/or multiphoton microscopy (optionally with
> slicing into confocal penetration depth compatible thickness), it can
> be brought to the lab and processed for sectioning,
> immunofluorescence, /in situ/ hybridization, H&E, IHC, etc.
>
> George
>
>
>
>
> At 12:30 PM 2/10/2008, you wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>> Search the CONFOCAL archive at
>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>> Hi Glen,
>>>
>>> Bob Zucker told me that if the tissue still has any water in it, the
>>> BABB will turn cloudy. Another item to be aware of with BABB is that
>>> it can solubilize some glues, so using it in an imaging chamber
>>> could results in leaks. Bob also emphasized that BABB is a very bad
>>> thing to have land on an objective lens or any other optic. One last
>>> item about BABB - I found some chemical company web sites that
>>> listed the refractive indices of several benzyl X compounds. Some
>>> are at or below RI=1.515. So, it may be feasible to find a mixture
>>> of two that is right at 1.515. That said, I'm excited about TDE
>>> because even neat, its RI is close to that of immersion oil and
>>> glass as to not matter except maybe hyper-critical deconvolution.
>>>
>>> I am currently storing the TDE (25 mL bottle, smallest that
>>> Sigma-Aldrich sells) in the chemical cabinet. I will probably
>>> aliquot it out into 0.5 or 1.0 mL eppendorf tubes to make it easier
>>> to distribute to users who want to try.
>>>
>>> My first user had ~5 um lung tissue sections. No way for me to tell
>>> whether they had shrinkage. There was an excellent article in one of
>>> the trade magazines (Microscopy Today?) a year or so on
>>> paraformaldehyde concentration in perfusion fixation. That, and
>>> other articles, lead me to expect that shrinkage comes from the PFA
>>> step, not from the later tissue clearing and dehydration.
>>> George
>>
>>
>> Hi George,
>>
>> On your last point: shrinkage
>>
>> In the late 1970s Alan Boyde and Elaine Maconnachie published a
>> number of papers (some noted below) describing experiments where they
>> followed the size of various specimens (fetal rat limb buds etc.)
>> during all the transitions between being alive and being an SEM
>> specimen (i.e., fixation, dehydration, replacement with intermediate
>> and transition liquids, actual drying). These images where then
>> analyzed quantitatively "on-line"  with a Quantimet image analyzing
>> computer.
>>
>> They used a wide variety of CDP protocols (different dehydrations,
>> different intermediate and transition liquids) as well as
>> freeze-drying from water ice and also from "ices" made by freezing
>> specimens immersed in non-polar liquids.
>>
>> Although every type of transition could produced changes in volume,
>> by far the largest changes were always associated with the final
>> steps of dehydration (say from 80% to 100% ethanol or acetone).
>>
>> I should also point out that the total shrinkage was "massive".  
>> Although freeze drying was best (about 40% VOLUME shrinkage as I
>> recall, most if it occurring  as the "last of the last" of the ice or
>> bound-water was removed as the specimen warmed up in the vacuum) ,
>> the best CPD protocols caused "only" about 60% VOLUME shrinkage.
>> Others produced 80% shrinkage.
>>
>> How could this be one might ask. Surely SEM specimens could not look
>> so "real" had they shrunk to this degree. Part of the answer has to
>> do with the fact that many SEM (and LM) specimens consist of cells
>> adhered to glass or plastic substrates. The glass does not shrink.
>> The cells attached to it become thinner, rather like the skin of an
>> animal pegged out on a frame and left in the sun to dry. As we seldom
>> take the trouble to measure their thickness, its reduction goes
>> unremarked.
>>
>> In Bob Bacallao's Chapter 18 in the Handbook, one can see images of
>> this effect. On the other hand, the limb buds were merely suspended
>> in liquid, not constrained in any way.  In addition, dehydration
>> protocols often extract lipids and there my be more of these in limb
>> buds than in the average tissue culture cell. Finally, cells with
>> high water content shrunk more.
>>
>> The basic message is the same: The proteins shrivel up as the last of
>> the water is removed, whether this occurs by by sublimation or by
>> dehydration.
>>
>> I very much suspect that the same will be true of specimens embedded
>> in clearing agents.
>>
>> Cheers,
>>
>> Jim P.
>>
>> Boyde, A., and Maconnachie, E., 1979, "Volume changes during preparation
>> of mouse embryonic tissue for scanning electron microscopy."/ Scanning
>> /2:149-163.
>> Boyde, A., and Maconnachie, E., 1981, "Morphological correlations with
>> dimensional change during SEM specimen preparation."/ Scanning Electron
>> Microsc./ 4:278-34.
>>
>>
>> --
>>    
>>               **********************************************
>> Prof. James B. Pawley,                                        Ph.  
>> 608-263-3147
>> Room 223, Zoology Research Building,                                
>> FAX  608-265-5315
>> 1117 Johnson Ave., Madison, WI, 53706                              
>> [hidden email]
>> 3D Microscopy of Living Cells Course, June 14-26, 2008, UBC,
>> Vancouver Canada
>> Info: http://www.3dcourse.ubc.ca/              Applications due by
>> March 15, 2008
>>                 "If it ain't diffraction, it must be statistics." Anon.
>
>
>
>
>  
>
> George McNamara, Ph.D.
> University of Miami, Miller School of Medicine
> Image Core
> Miami, FL 33010
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://home.earthlink.net/~pubspectra/ 
> <http://home.earthlink.net/%7Epubspectra/>
> http://home.earthlink.net/~geomcnamara/ 
> <http://home.earthlink.net/%7Egeomcnamara/>
> http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc 
> (Analytical Imaging Core Facility)
>
>
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: penetration depth

In reply to this post by George McNamara
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi,
Ok, so it is  a dehydration issue.  Incomplete dehydration of tissue  
passing through xylene will also cause Permount and DPX to become  
cloudy, a few days following coverslipping.  I've been rigorous on  
dehydration and passing through graded mixtures of absolution EtOH  
and the MSBB mixture without cloudiness.  Xylene tends to make  
tissues brittle.  Yeah, the BB mixtures dissolve almost anything,  
even makes temporary wells of  silicone grease very runny.  I've been  
mounting in aluminum frames used to support laser microdissection  
films to which I've attached a measured  24x60 mm coverglass using  
silicone aquarium glue.  So far, they've survived a year or so.

I wonder if storing TDE at  in aliquots at -80 would avoid the need  
to fill the bottle with inert gas after opening.   The class of  
compounds to which TDE belongs is supposed to have some higher RI  
members, which would be helpful for high RI samples.

thanks,
Glen

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Glen,
>
> Bob Zucker told me that if the tissue still has any water in it,  
> the BABB will turn cloudy. Another item to be aware of with BABB is  
> that it can solubilize some glues, so using it in an imaging  
> chamber could results in leaks. Bob also emphasized that BABB is a  
> very bad thing to have land on an objective lens or any other  
> optic. One last item about BABB - I found some chemical company web  
> sites that listed the refractive indices of several benzyl X  
> compounds. Some are at or below RI=1.515. So, it may be feasible to  
> find a mixture of two that is right at 1.515. That said, I'm  
> excited about TDE because even neat, its RI is close to that of  
> immersion oil and glass as to not matter except maybe hyper-
> critical deconvolution.
>
> I am currently storing the TDE (25 mL bottle, smallest that Sigma-
> Aldrich sells) in the chemical cabinet. I will probably aliquot it  
> out into 0.5 or 1.0 mL eppendorf tubes to make it easier to  
> distribute to users who want to try.
>
> My first user had ~5 um lung tissue sections. No way for me to tell  
> whether they had shrinkage. There was an excellent article in one  
> of the trade magazines (Microscopy Today?) a year or so on  
> paraformaldehyde concentration in perfusion fixation. That, and  
> other articles, lead me to expect that shrinkage comes from the PFA  
> step, not from the later tissue clearing and dehydration.
>
> George
>
>
>
> At 11:25 AM 2/8/2008, you wrote:
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Dear George,
>> Did Bob say why the xylene was important?  I would  expect that BABB
>> would be as alcohol-soluble as methyl salicylate:benzyl benzoate.
>> Maybe the xylene was ensuring the last of the water is out, but
>> thorough dehydration should be sufficient.  Many mixtures of solvents
>> are given in this book: Spalteholz, W., 1914. Über das
>> Durchsichtigmachen von menschlichen und tierischen Präparaten und
>> seine theoretischen Bedingungen. Hirzel, Leipzig.
>>
>> Glad to hear someone else has tried TDE. Do you notice any tissue
>> shrinkage? How are you storing it?  I've haven't yet opened my
>> bottle.  One post doc is using my mixture of MSBB on 1200 µm brain
>> slices, but I'm going to try TDE on brain slices <400 µm with grad
>> student. Thinner slices tend to curl after dehydration.
>>
>> Regards,
>> Glen
>>
>> Glen MacDonald
>> Core for Communication Research
>> Virginia Merrill Bloedel Hearing Research Center
>> Box 357923
>> University of Washington
>> Seattle, WA 98195-7923  USA
>> (206) 616-4156
>> [hidden email]
>>
>> *********************************************************************
>> *** ******
>> The box said "Requires WindowsXP or better", so I bought a Macintosh.
>> *********************************************************************
>> *** ******
>>
>>
>> On Feb 7, 2008, at 6:58 PM, George McNamara wrote:
>>
>>> Search the CONFOCAL archive at
>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>
>>> Hi Judy,
>>>
>>> I was impressed with 2,2'-thiodiethanol, used neat (R.I. 1.52;
>>> Sigma-Aldrich). Completely black background (N=2 slides, one
>>> confocal session). See Staudt, Hell et al 2007 Microsc Res Tech for
>>> original reference. They used RI 1.515 by adding 3% H2O. I figured
>>> no reason to pollute the TDE with any additives. One oddity was
>>> that by eye, both blood vessel elastin fiber autofluorescence was
>>> decreased (488 nm excitation, green emission) and the DAPI nuclear
>>> counterstain was quenched. Both came through nicely in the confocal.
>>>
>>> Robert Zucker has several confocal papers on BABB - benzyl  
>>> alcohol/ benzyl benzoate as a clearing agent (see another  
>>> responder's
>>> message for other names). Bob told me the key is to completely
>>> dehydrate the specimen through EtOH steps into xylene, before going
>>> to BABB. See his papers for exact steps.
>>>
>>> best wishes,
>>>
>>> George
>>>
>>>
>>> At 11:09 AM 2/5/2008, you wrote:
>>>> Search the CONFOCAL archive at
>>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>>
>>>> What's the best method of making a sample more transparent to
>>>> reduce scattering, without changing its chemistry too much? (other
>>>> than methyl salicylate and glycerol).
>>>>
>>>> Thanks.
>>>> Judy
>>>>
>>>> Judy Trogadis
>>>> Bio-Imaging Coordinator
>>>> St. Michael's Hospital, 7Queen
>>>> 30 Bond St.
>>>> Toronto, ON M5B 1W8, Canada
>>>> ph:  416-864-6060  x6337
>>>> pager: 416-685-9219
>>>> fax: 416-864-6043
>>>> [hidden email]
>>>>
>>>>
>>>> >>> Bill Miller <[hidden email]> 02/04/08 9:16 PM >>>
>>>> Search the CONFOCAL archive at
>>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>>
>>>> ultimately the lower NA with its longer working distance wins if  
>>>> the
>>>> sample is clear enough..
>>>>
>>>>
>>>> At 08:31 PM 2/4/2008 -0500, you wrote:
>>>> >Search the CONFOCAL archive at
>>>> >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>>> >Hello all!
>>>> >
>>>> >I was wondering for which one penetration depth is higher: NA:1.2
>>>> >60x lens or NA : 0.3 10x lens?
>>>> >
>>>> >thanks
>>>> >Sarah
>>>
>>>
>>>
>>>
>>>
>>>
>>> George McNamara, Ph.D.
>>> University of Miami, Miller School of Medicine
>>> Image Core
>>> Miami, FL 33010
>>> [hidden email]
>>> [hidden email]
>>> 305-243-8436 office
>>> http://home.earthlink.net/~pubspectra/
>>> http://home.earthlink.net/~geomcnamara/
>>> http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc
>>> (Analytical Imaging Core Facility)
>
>
>
>
>
>
> George McNamara, Ph.D.
> University of Miami, Miller School of Medicine
> Image Core
> Miami, FL 33010
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://home.earthlink.net/~pubspectra/
> http://home.earthlink.net/~geomcnamara/
> http://www.sylvester.org/research/SR_lab_analytical.asp?ana=desc 
> (Analytical Imaging Core Facility)