To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, I have a user who is assessing bacterial biofilm clearance by ciliated airway epithelial cells grown on air liquid interface cultures. He then fixes the cultures, cuts out and divides the membrane and does FISH for the bacteria, immuno for the cilia plus nuclear counterstain. The filter slices are mounted in mountant between 2 coverslips and imaged on SP5 in tile scan mode, imaging a long, thin radial strip of approx 45x1 fields at 512x512 resolution and multiple Z levels (up to 50ish) It is impossible to get the filters perfectly flat - they tilt and buckle so we need to set a large Z range to encompass absolute highest and lowest points and many Z slices. Ultimately he wants to calculate an stimate of biofilm volume My 2 problems, seeking advice and solutions: (1) because of the wide Z range, some slices on some fields cut into the filter (autofluorescence and non specific binding at holes) or close to coverslip (flash). Because of the undulations of the membrane, this varies from field of view to field of view. Therefore I need a way to crop or blank Z slices off the data set in a manner which is field of view specific, rather than global, but still retains the relative Z position in the overall stack. (2) Despite the fields forming an absolute linear strip, Leica's autostitcher sometimes makes a complete hash of it. Is there any easy way to get a stitch of Z stacks based simply on relative location. We are talking reasonably large datafiles here (several GB) Thanks in Advance, Dave Johnston, Biomedical Imaging Unit, Southampton, UK. PS, learned the hard way, if doing tile sets, make sure scan head rotation is set to 0 :-) |
![]() ![]() |
G. Esteban Fernandez |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Have you tried ImageJ for stitching? Fiji ImageJ (http://fiji.sc/) comes pre-loaded with s stitching plugin where you can specify the relative location of tiles to prime the autostitching algorithm (Plugins > Stitching > Grid/Collection stitching). It has worked very well for me, the priming made a big difference and even stitched some low-contrast brightfield images well. -Esteban On Thu, Aug 16, 2012 at 12:30 AM, David Johnston <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > I have a user who is assessing bacterial biofilm clearance by ciliated airway > epithelial cells grown on air liquid interface cultures. He then fixes the cultures, > cuts out and divides the membrane and does FISH for the bacteria, immuno for > the cilia plus nuclear counterstain. The filter slices are mounted in mountant > between 2 coverslips and imaged on SP5 in tile scan mode, imaging a long, thin > radial strip of approx 45x1 fields at 512x512 resolution and multiple Z levels (up > to 50ish) It is impossible to get the filters perfectly flat - they tilt and buckle > so we need to set a large Z range to encompass absolute highest and lowest > points and many Z slices. Ultimately he wants to calculate an stimate of > biofilm volume > > My 2 problems, seeking advice and solutions: > (1) because of the wide Z range, some slices on some fields cut into the filter > (autofluorescence and non specific binding at holes) or close to coverslip > (flash). Because of the undulations of the membrane, this varies from field of > view to field of view. Therefore I need a way to crop or blank Z slices off the > data set in a manner which is field of view specific, rather than global, but still > retains the relative Z position in the overall stack. > > (2) Despite the fields forming an absolute linear strip, Leica's autostitcher > sometimes makes a complete hash of it. Is there any easy way to get a stitch > of Z stacks based simply on relative location. > > We are talking reasonably large datafiles here (several GB) > > Thanks in Advance, > Dave Johnston, > Biomedical Imaging Unit, Southampton, UK. > > PS, learned the hard way, if doing tile sets, make sure scan head rotation is > set to 0 :-) |
![]() ![]() |
George McNamara |
![]() |
In reply to this post by daj1u06
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Dave, Contact your Leica confocal sales rep and have them visit to demo the Leica MATRIX module(s) to do what you need. They might also want to bring with them a Leica MATRIX application(s) expert. A quote for a new 64-bit PC might also be useful. I've already complained to Leica that their LAS AF stage tiling acquisition and image stitching is bad (I probably complained about it being bad in TCS and LCS when I managed an SP1 - accomplished nothing). Not that Zeiss LSM710 / ZEN software is any better. George On 8/16/2012 3:30 AM, David Johnston wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > I have a user who is assessing bacterial biofilm clearance by ciliated airway > epithelial cells grown on air liquid interface cultures. He then fixes the cultures, > cuts out and divides the membrane and does FISH for the bacteria, immuno for > the cilia plus nuclear counterstain. The filter slices are mounted in mountant > between 2 coverslips and imaged on SP5 in tile scan mode, imaging a long, thin > radial strip of approx 45x1 fields at 512x512 resolution and multiple Z levels (up > to 50ish) It is impossible to get the filters perfectly flat - they tilt and buckle > so we need to set a large Z range to encompass absolute highest and lowest > points and many Z slices. Ultimately he wants to calculate an stimate of > biofilm volume > > My 2 problems, seeking advice and solutions: > (1) because of the wide Z range, some slices on some fields cut into the filter > (autofluorescence and non specific binding at holes) or close to coverslip > (flash). Because of the undulations of the membrane, this varies from field of > view to field of view. Therefore I need a way to crop or blank Z slices off the > data set in a manner which is field of view specific, rather than global, but still > retains the relative Z position in the overall stack. > > (2) Despite the fields forming an absolute linear strip, Leica's autostitcher > sometimes makes a complete hash of it. Is there any easy way to get a stitch > of Z stacks based simply on relative location. > > We are talking reasonably large datafiles here (several GB) > > Thanks in Advance, > Dave Johnston, > Biomedical Imaging Unit, Southampton, UK. > > PS, learned the hard way, if doing tile sets, make sure scan head rotation is > set to 0 :-) > > |
![]() ![]() |
Barlow, Andrew |
![]() |
In reply to this post by daj1u06
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Dave, We've recently implemented stitching of 3D, time resolved datasets in Volocity. The relative positions of tiles is read from Leica lif files, which greatly speeds the process up (compared to blind stitching). We've had some great feedback about this feature, please contact me offline if you'd like a demonstration. Regards, Andrew JOIN US FOR OUR VOLOCITY AND COLUMBUS USER GROUP MEETINGS AT THE EUROPEAN MICROSCOPY CONGRESS 2012 Manchester, UK, Friday 21st September 2012, 2pm - 6pm Click here to register Andrew Barlow PhD | Market Development Leader, CI & A PerkinElmer | For the Better [hidden email] Phone: + 44 2476 692229| Fax: +44 2476 690091 | Mobile: +44 7799 795 999 Millburn Hill Road, Coventry, CV4 7HS, UK www.perkinelmer.com Please consider the environment before printing this e-mail. This e-mail message and any attachments are confidential and proprietary to PerkinElmer, Inc. If you are not the intended recipient of this message, please inform the sender by replying to this email or sending a message to the sender and destroy the message and any attachments. Thank you. > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of David Johnston > Sent: 16 August 2012 08:30 > To: [hidden email] > Subject: processing and Z cropping tiled Leica images > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > > I have a user who is assessing bacterial biofilm clearance by ciliated > airway epithelial cells grown on air liquid interface cultures. He then > fixes the cultures, cuts out and divides the membrane and does FISH for > the bacteria, immuno for the cilia plus nuclear counterstain. The > filter slices are mounted in mountant between 2 coverslips and imaged > on SP5 in tile scan mode, imaging a long, thin radial strip of approx > 45x1 fields at 512x512 resolution and multiple Z levels (up to 50ish) > It is impossible to get the filters perfectly flat - they tilt and > buckle so we need to set a large Z range to encompass absolute highest > and lowest points and many Z slices. Ultimately he wants to calculate > an stimate of biofilm volume > > My 2 problems, seeking advice and solutions: > (1) because of the wide Z range, some slices on some fields cut into > the filter (autofluorescence and non specific binding at holes) or > close to coverslip (flash). Because of the undulations of the membrane, > this varies from field of view to field of view. Therefore I need a way > to crop or blank Z slices off the data set in a manner which is field > of view specific, rather than global, but still retains the relative Z > position in the overall stack. > > (2) Despite the fields forming an absolute linear strip, Leica's > autostitcher sometimes makes a complete hash of it. Is there any easy > way to get a stitch of Z stacks based simply on relative location. > > We are talking reasonably large datafiles here (several GB) > > Thanks in Advance, > Dave Johnston, > Biomedical Imaging Unit, Southampton, UK. > > PS, learned the hard way, if doing tile sets, make sure scan head > rotation is set to 0 :-) |
![]() ![]() |
Thomas Trusk |
![]() |
In reply to this post by daj1u06
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I've been experimenting quite a bit with the latest incarnation of Leica's stitching routine in LAS, and it is MUCH improved. The automatic mode does use content as a guide, so blank or sparsely populated images present a problem. If you turn AUTO off, the stitching is faster and edge-to-edge. I should note that Leica offers a new option, that I have not had a chance to see, that supposedly pre-maps focus over the entire area to be stitched. I've been told this function would allow one to scan only areas where regions of interest exist, preventing the blank image problem and saving lots of storage space. Thomas C. Trusk, PhD Director, Josh Spruill Imaging Facility Regenerative Medicine and Cell Biology Medical University of South Carolina Charleston, SC 29425 http://mmi.musc.edu |