question about laser

Hi everyone,

We are looking for a coherent (or continuous wave) laser ranging from  
380nm to 400nm. So far we haven't found anything impressive. Does  
anyone know any laser between that range? Thanks a lot!

Yue
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

A CW laser is by definition narrowband; there's no way you can have one with 20nm spectral bandwidth.  The only thing I can think of off hand that would give you the spectral range you are looking for is an LED.  The only other alternative would be something like a supercontinuum (or so-called 'white' laser) but that emits over a huge range and would be overkill.  I'd recommend trying to find a really powerful LED in that range, or even a bunch of lower-powered LEDs bundled into an array.

Craig


On Wed, Mar 4, 2009 at 10:43 AM, <[hidden email]> wrote:
Hi everyone,

We are looking for a coherent (or continuous wave) laser ranging from 380nm to 400nm. So far we haven't found anything impressive. Does anyone know any laser between that range? Thanks a lot!

Yue

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

Ah, just realized I misunderstood your question.  You just need a laser line between 380 and 400.  Your best bet there would be to get a high-powered IR diode laser and use a crystal to double it.  A diode laser at 780nm doubled would give you 390nm.

Craig


On Wed, Mar 4, 2009 at 11:49 AM, Craig Brideau <[hidden email]> wrote:
A CW laser is by definition narrowband; there's no way you can have one with 20nm spectral bandwidth.  The only thing I can think of off hand that would give you the spectral range you are looking for is an LED.  The only other alternative would be something like a supercontinuum (or so-called 'white' laser) but that emits over a huge range and would be overkill.  I'd recommend trying to find a really powerful LED in that range, or even a bunch of lower-powered LEDs bundled into an array.

Craig



On Wed, Mar 4, 2009 at 10:43 AM, <[hidden email]> wrote:
Hi everyone,

We are looking for a coherent (or continuous wave) laser ranging from 380nm to 400nm. So far we haven't found anything impressive. Does anyone know any laser between that range? Thanks a lot!

Yue


Mario-2 Mario-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

Re: question about laser
Speaking of which, whatever happened to the promised solid state laser that is basically the same as the 405 nm but tuned to emit as short as 385 nm? I may be a bit off, but the DAPI fluorescence emission is about 7 times higher at 385 nm (~45% max) compared to 405 nm (~6% max).

Mario

Ah, just realized I misunderstood your question.  You just need a laser line between 380 and 400.  Your best bet there would be to get a high-powered IR diode laser and use a crystal to double it.  A diode laser at 780nm doubled would give you 390nm.

Craig
On Wed, Mar 4, 2009 at 11:49 AM, Craig Brideau <[hidden email]> wrote:
A CW laser is by definition narrowband; there's no way you can have one with 20nm spectral bandwidth.  The only thing I can think of off hand that would give you the spectral range you are looking for is an LED.  The only other alternative would be something like a supercontinuum (or so-called 'white' laser) but that emits over a huge range and would be overkill.  I'd recommend trying to find a really powerful LED in that range, or even a bunch of lower-powered LEDs bundled into an array.

Craig



On Wed, Mar 4, 2009 at 10:43 AM, <[hidden email]> wrote:
Hi everyone,

We are looking for a coherent (or continuous wave) laser ranging from 380nm to 400nm. So far we haven't found anything impressive. Does anyone know any laser between that range? Thanks a lot!

Yue


-- 
________________________________________________________________________________
Mario M. Moronne, Ph.D.

[hidden email]
[hidden email]
[hidden email]
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

Sorry for the confusion. I wasn't clear enough about my question. You  
are right, Craig. I'm looking for a laser line that's between 380nm  
and 400nm.

Thank you all for the good suggestions!

Yue


Quoting Mario <[hidden email]>:

> Speaking of which, whatever happened to the promised solid state laser
> that is basically the same as the 405 nm but tuned to emit as short as
> 385 nm? I may be a bit off, but the DAPI fluorescence emission is about
> 7 times higher at 385 nm (~45% max) compared to 405 nm (~6% max).
>
> Mario
>
>> Ah, just realized I misunderstood your question.  You just need a  
>> laser line between 380 and 400.  Your best bet there would be to  
>> get a high-powered IR diode laser and use a crystal to double it.    
>> A diode laser at 780nm doubled would give you 390nm.
>>
>> Craig
>>
>> On Wed, Mar 4, 2009 at 11:49 AM, Craig Brideau  
>> <<mailto:[hidden email]>[hidden email]> wrote:
>>
>> A CW laser is by definition narrowband; there's no way you can have  
>>  one with 20nm spectral bandwidth.  The only thing I can think of  
>> off hand that would give you the spectral range you are looking for  
>>  is an LED.  The only other alternative would be something like a  
>> supercontinuum (or so-called 'white' laser) but that emits over a  
>> huge range and would be overkill.  I'd recommend trying to find a  
>> really powerful LED in that range, or even a bunch of lower-powered  
>>  LEDs bundled into an array.
>>
>> Craig
>>
>>
>>
>> On Wed, Mar 4, 2009 at 10:43 AM,  
>> <<mailto:[hidden email]>[hidden email]> wrote:
>>
>> Hi everyone,
>>
>> We are looking for a coherent (or continuous wave) laser ranging  
>> from 380nm to 400nm. So far we haven't found anything impressive.  
>> Does anyone know any laser between that range? Thanks a lot!
>>
>> Yue
>
>
> --
> ________________________________________________________________________________
> Mario M. Moronne, Ph.D.
>
> [hidden email]
> [hidden email]
> [hidden email]
femto@sciner.com femto@sciner.com
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

In reply to this post by yue_li
Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.

Best regards,
Sergey Egorov
Del Mar Photonics
4119 Twilight Ridge
San Diego, CA 92130
tel (858) 876-3133
fax (858) 630-2376
http://www.dmphotonics.com/


[hidden email] wrote:
> Hi everyone,
>
> We are looking for a coherent (or continuous wave) laser ranging from
> 380nm to 400nm. So far we haven't found anything impressive. Does
> anyone know any laser between that range? Thanks a lot!
>
> Yue
>  
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

In reply to this post by yue_li
Picoquant supply diode lasers (pulsed or cw) from 375nm
to something like 470nm (you can get whatever you want.)
I have no  connection with the company (except as a
satisfied customer).  I'm sure there are other suppliers
as well.  They are quite cheap (a tiny fraction of a
frequency doubled TiS).  You do have to think a bit about
your optics at this wavelength (and how you align a beam
you can't see).  

                                             Guy


Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
   http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861       http://www.guycox.net
______________________________________________
 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of [hidden email]
Sent: Thursday, 5 March 2009 10:48 AM
To: [hidden email]
Subject: Re: question about laser

Sorry for the confusion. I wasn't clear enough about my question. You  
are right, Craig. I'm looking for a laser line that's between 380nm  
and 400nm.

Thank you all for the good suggestions!

Yue


Quoting Mario <[hidden email]>:

> Speaking of which, whatever happened to the promised solid state laser
> that is basically the same as the 405 nm but tuned to emit as short as
> 385 nm? I may be a bit off, but the DAPI fluorescence emission is
about

> 7 times higher at 385 nm (~45% max) compared to 405 nm (~6% max).
>
> Mario
>
>> Ah, just realized I misunderstood your question.  You just need a  
>> laser line between 380 and 400.  Your best bet there would be to  
>> get a high-powered IR diode laser and use a crystal to double it.    
>> A diode laser at 780nm doubled would give you 390nm.
>>
>> Craig
>>
>> On Wed, Mar 4, 2009 at 11:49 AM, Craig Brideau  
>> <<mailto:[hidden email]>[hidden email]> wrote:
>>
>> A CW laser is by definition narrowband; there's no way you can have  
>>  one with 20nm spectral bandwidth.  The only thing I can think of  
>> off hand that would give you the spectral range you are looking for  
>>  is an LED.  The only other alternative would be something like a  
>> supercontinuum (or so-called 'white' laser) but that emits over a  
>> huge range and would be overkill.  I'd recommend trying to find a  
>> really powerful LED in that range, or even a bunch of lower-powered  
>>  LEDs bundled into an array.
>>
>> Craig
>>
>>
>>
>> On Wed, Mar 4, 2009 at 10:43 AM,  
>> <<mailto:[hidden email]>[hidden email]> wrote:
>>
>> Hi everyone,
>>
>> We are looking for a coherent (or continuous wave) laser ranging  
>> from 380nm to 400nm. So far we haven't found anything impressive.  
>> Does anyone know any laser between that range? Thanks a lot!
>>
>> Yue
>
>
> --
>
________________________________________________________________________
________
> Mario M. Moronne, Ph.D.
>
> [hidden email]
> [hidden email]
> [hidden email]

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:
4/03/2009 7:41 AM
 
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

Thank you so much for the information. Cost is a big factor in our  
decision. I'll check that company out. We can't align the laser in the  
lab, but we have someone else help us to do it.

Yue


Quoting Guy Cox <[hidden email]>:

> Picoquant supply diode lasers (pulsed or cw) from 375nm
> to something like 470nm (you can get whatever you want.)
> I have no  connection with the company (except as a
> satisfied customer).  I'm sure there are other suppliers
> as well.  They are quite cheap (a tiny fraction of a
> frequency doubled TiS).  You do have to think a bit about
> your optics at this wavelength (and how you align a beam
> you can't see).
>
>                                              Guy
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>    http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861       http://www.guycox.net
> ______________________________________________
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of [hidden email]
> Sent: Thursday, 5 March 2009 10:48 AM
> To: [hidden email]
> Subject: Re: question about laser
>
> Sorry for the confusion. I wasn't clear enough about my question. You
> are right, Craig. I'm looking for a laser line that's between 380nm
> and 400nm.
>
> Thank you all for the good suggestions!
>
> Yue
>
>
> Quoting Mario <[hidden email]>:
>
>> Speaking of which, whatever happened to the promised solid state laser
>> that is basically the same as the 405 nm but tuned to emit as short as
>> 385 nm? I may be a bit off, but the DAPI fluorescence emission is
> about
>> 7 times higher at 385 nm (~45% max) compared to 405 nm (~6% max).
>>
>> Mario
>>
>>> Ah, just realized I misunderstood your question.  You just need a
>>> laser line between 380 and 400.  Your best bet there would be to
>>> get a high-powered IR diode laser and use a crystal to double it.
>>> A diode laser at 780nm doubled would give you 390nm.
>>>
>>> Craig
>>>
>>> On Wed, Mar 4, 2009 at 11:49 AM, Craig Brideau
>>> <<mailto:[hidden email]>[hidden email]> wrote:
>>>
>>> A CW laser is by definition narrowband; there's no way you can have
>>>  one with 20nm spectral bandwidth.  The only thing I can think of
>>> off hand that would give you the spectral range you are looking for
>>>  is an LED.  The only other alternative would be something like a
>>> supercontinuum (or so-called 'white' laser) but that emits over a
>>> huge range and would be overkill.  I'd recommend trying to find a
>>> really powerful LED in that range, or even a bunch of lower-powered
>>>  LEDs bundled into an array.
>>>
>>> Craig
>>>
>>>
>>>
>>> On Wed, Mar 4, 2009 at 10:43 AM,
>>> <<mailto:[hidden email]>[hidden email]> wrote:
>>>
>>> Hi everyone,
>>>
>>> We are looking for a coherent (or continuous wave) laser ranging
>>> from 380nm to 400nm. So far we haven't found anything impressive.
>>> Does anyone know any laser between that range? Thanks a lot!
>>>
>>> Yue
>>
>>
>> --
>>
> ________________________________________________________________________
> ________
>> Mario M. Moronne, Ph.D.
>>
>> [hidden email]
>> [hidden email]
>> [hidden email]
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:
> 4/03/2009 7:41 AM
>
>
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

In reply to this post by femto@sciner.com
I don't quite understand. The laser beam we are looking for is between  
380 and 400nm. Are you thinking of the doubler? Could you please give  
more detail? Thanks a lot.

Yue


Quoting "Femto @Sciner" <[hidden email]>:

> Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.
>
> Best regards,
> Sergey Egorov
> Del Mar Photonics
> 4119 Twilight Ridge
> San Diego, CA 92130
> tel (858) 876-3133
> fax (858) 630-2376
> http://www.dmphotonics.com/
>
>
> [hidden email] wrote:
>> Hi everyone,
>>
>> We are looking for a coherent (or continuous wave) laser ranging  
>> from 380nm to 400nm. So far we haven't found anything impressive.  
>> Does anyone know any laser between that range? Thanks a lot!
>>
>> Yue
>>
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

Yes, he was referring to a doubler.  The problem is people are
still not clear whether you want to tune the wavelength between
380 and 400 nm  (in which case Sergey's suggestion is the most
practical way to go, but is very expensive) or whether you just
want one fixed wavelength within that range, which is easy and
cheap to get with a diode.

                                                  Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Friday, 6 March 2009 4:18 AM
To: [hidden email]
Subject: Re: question about laser

I don't quite understand. The laser beam we are looking for is between 380 and 400nm. Are you thinking of the doubler? Could you please give more detail? Thanks a lot.

Yue


Quoting "Femto @Sciner" <[hidden email]>:

> Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.
>
> Best regards,
> Sergey Egorov
> Del Mar Photonics
> 4119 Twilight Ridge
> San Diego, CA 92130
> tel (858) 876-3133
> fax (858) 630-2376
> http://www.dmphotonics.com/
>
>
> [hidden email] wrote:
>> Hi everyone,
>>
>> We are looking for a coherent (or continuous wave) laser ranging  
>> from 380nm to 400nm. So far we haven't found anything impressive.  
>> Does anyone know any laser between that range? Thanks a lot!
>>
>> Yue
>>

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date: 4/03/2009 7:41 AM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date: 4/03/2009 7:41 AM
 
yue_li yue_li
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

I just want one fixed wavelength within that range eventually. The  
problem is that I don't know which one will work the best on my  
sample. I only know it should lay between 380 and 400nm from my other  
experiments. And I'm hoping to find an inexpensive way to find out.

Yue


Quoting Guy Cox <[hidden email]>:

> Yes, he was referring to a doubler.  The problem is people are
> still not clear whether you want to tune the wavelength between
> 380 and 400 nm  (in which case Sergey's suggestion is the most
> practical way to go, but is very expensive) or whether you just
> want one fixed wavelength within that range, which is easy and
> cheap to get with a diode.
>
>                                                   Guy
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On Behalf Of  
> [hidden email]
> Sent: Friday, 6 March 2009 4:18 AM
> To: [hidden email]
> Subject: Re: question about laser
>
> I don't quite understand. The laser beam we are looking for is  
> between 380 and 400nm. Are you thinking of the doubler? Could you  
> please give more detail? Thanks a lot.
>
> Yue
>
>
> Quoting "Femto @Sciner" <[hidden email]>:
>
>> Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.
>>
>> Best regards,
>> Sergey Egorov
>> Del Mar Photonics
>> 4119 Twilight Ridge
>> San Diego, CA 92130
>> tel (858) 876-3133
>> fax (858) 630-2376
>> http://www.dmphotonics.com/
>>
>>
>> [hidden email] wrote:
>>> Hi everyone,
>>>
>>> We are looking for a coherent (or continuous wave) laser ranging
>>> from 380nm to 400nm. So far we haven't found anything impressive.
>>> Does anyone know any laser between that range? Thanks a lot!
>>>
>>> Yue
>>>
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:  
> 4/03/2009 7:41 AM
>
>
> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:  
> 4/03/2009 7:41 AM
>
>
Sam's Mail Sam's Mail
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser. .

Yue,

Please, feel free to tell us a little bit more details than you currently are providing.  There is a wealth of information on this list, and with more information from you comes more knowledge from the group.  What is your application?  It is potentially a cheaper and more practical solution to buy the best laser you can afford that fits the needs of your fluorophore, and then purchase custom dichroic mirrors and bandpass filters if necessary.  These can be more affordable to buy than new lasers, and can be custom manufactured to meet your experimental needs from a few companies.  Chroma provides this service in a timely fashion. (no commercial interest)

--
Samuel A. Connell
Director of Light Microscopy
Cell & Tissue Imaging Center
St. Jude Children's Research Hospital
262 Danny Thomas Place
Memphis, TN 38105-3678
Office (901) 495-2536
Cell (901) 603-3162
[hidden email]



On 3/5/09 9:47 PM, "[hidden email]" <[hidden email]> wrote:

I just want one fixed wavelength within that range eventually. The
problem is that I don't know which one will work the best on my
sample. I only know it should lay between 380 and 400nm from my other
experiments. And I'm hoping to find an inexpensive way to find out.

Yue


Quoting Guy Cox <[hidden email]>:

> Yes, he was referring to a doubler.  The problem is people are
> still not clear whether you want to tune the wavelength between
> 380 and 400 nm  (in which case Sergey's suggestion is the most
> practical way to go, but is very expensive) or whether you just
> want one fixed wavelength within that range, which is easy and
> cheap to get with a diode.
>
>                                                   Guy
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of
> [hidden email]
> Sent: Friday, 6 March 2009 4:18 AM
> To: [hidden email]
> Subject: Re: question about laser
>
> I don't quite understand. The laser beam we are looking for is
> between 380 and 400nm. Are you thinking of the doubler? Could you
> please give more detail? Thanks a lot.
>
> Yue
>
>
> Quoting "Femto @Sciner" <[hidden email]>:
>
>> Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.
>>
>> Best regards,
>> Sergey Egorov
>> Del Mar Photonics
>> 4119 Twilight Ridge
>> San Diego, CA 92130
>> tel (858) 876-3133
>> fax (858) 630-2376
>> http://www.dmphotonics.com/
>>
>>
>> [hidden email] wrote:
>>> Hi everyone,
>>>
>>> We are looking for a coherent (or continuous wave) laser ranging
>>> from 380nm to 400nm. So far we haven't found anything impressive.
>>> Does anyone know any laser between that range? Thanks a lot!
>>>
>>> Yue
>>>
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:
> 4/03/2009 7:41 AM
>
>
> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:
> 4/03/2009 7:41 AM
>
>



________________________________
Email Disclaimer: www.stjude.org/emaildisclaimer
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: question about laser

In reply to this post by yue_li
Can't you just check your fluorochrome on a spectrofluorimeter?

If for some reason you actually need to form images, why not
use a tunable filter with a mercury or metal halide lamp?
(Or even just buy a few different excitation filters if a
tunable filter is too dear.)

                                              Guy



Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
    http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
     http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Friday, 6 March 2009 2:48 PM
To: [hidden email]
Subject: Re: question about laser

I just want one fixed wavelength within that range eventually. The problem is that I don't know which one will work the best on my sample. I only know it should lay between 380 and 400nm from my other experiments. And I'm hoping to find an inexpensive way to find out.

Yue


Quoting Guy Cox <[hidden email]>:

> Yes, he was referring to a doubler.  The problem is people are still
> not clear whether you want to tune the wavelength between 380 and 400
> nm  (in which case Sergey's suggestion is the most practical way to
> go, but is very expensive) or whether you just want one fixed
> wavelength within that range, which is easy and cheap to get with a
> diode.
>
>                                                   Guy
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>     http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit,
> Madsen Building F09, University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>      http://www.guycox.net
> -----Original Message-----
> From: Confocal Microscopy List  
> [mailto:[hidden email]] On Behalf Of  
> [hidden email]
> Sent: Friday, 6 March 2009 4:18 AM
> To: [hidden email]
> Subject: Re: question about laser
>
> I don't quite understand. The laser beam we are looking for is  
> between 380 and 400nm. Are you thinking of the doubler? Could you  
> please give more detail? Thanks a lot.
>
> Yue
>
>
> Quoting "Femto @Sciner" <[hidden email]>:
>
>> Second Harmonic of CW Ti:Sapphire tunable between 760 and 800 nm.
>>
>> Best regards,
>> Sergey Egorov
>> Del Mar Photonics
>> 4119 Twilight Ridge
>> San Diego, CA 92130
>> tel (858) 876-3133
>> fax (858) 630-2376
>> http://www.dmphotonics.com/
>>
>>
>> [hidden email] wrote:
>>> Hi everyone,
>>>
>>> We are looking for a coherent (or continuous wave) laser ranging
>>> from 380nm to 400nm. So far we haven't found anything impressive.
>>> Does anyone know any laser between that range? Thanks a lot!
>>>
>>> Yue
>>>
>
> No virus found in this incoming message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:  
> 4/03/2009 7:41 AM
>
>
> No virus found in this outgoing message.
> Checked by AVG.
> Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date:  
> 4/03/2009 7:41 AM
>
>

No virus found in this incoming message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date: 4/03/2009 7:41 AM
 

No virus found in this outgoing message.
Checked by AVG.
Version: 7.5.557 / Virus Database: 270.11.7/1983 - Release Date: 4/03/2009 7:41 AM