		Dear colleagues,
 we are looking for suitable illumination & camera/s for our Leica Z6 MacroFluo. It is used for both fluorescence and absorption contrast samples. Has anybody out there experience with LED illumination in combination with a monochrome camera to generate 'true color' images? Your advice is welcome, commercial replies off list, please. Thank you, jens --- dr. jens rietdorf head microscopy facility novartis research foundation friedrich-miescher-institute, wro1066.2.16 maulbeerstr.66, 4058 basel, switzerland txt: rietdorf at fmi dot ch fon: +41.61.69.75172 fax: +41.61.69.73976 | 
|    | Peter Gabriel Pitrone | 
|   | 
		Hello Jens,
 You would need a large amount of pixels to digitally resolve your images with such low magnification, and a large chip to collect light from fluorescence. That is a difficult thing to find. One interesting color imaging technique is made by a company called Foveon: http:// www.foveon.com/ They make a 14 megapixel 3 layer CMOS chip (called Foveon X3) that is 21x14 mm, that might be useful to you in this regard. Here is a link to the PDF: http://www.foveon.com/files/ F13_image_sensor_Product_Flier_RevD.pdf I have no affiliation with the company, and have not even tried it out, but am intrigued by the technology. You could also get a normal monochrome CCD Camera with an electronic LCD trigger, like Diagnostic Instruments SPOT RT Slider from Visitron GmbH or QImaging Regina with a trigger from Chromaphor. Same disclaimer as above. Good luck on your search! Pete On Oct 28, 2009, at 2:23 PM, Rietdorf, Jens wrote: Dear colleagues, we are looking for suitable illumination & camera/s for our Leica Z6 MacroFluo. It is used for both fluorescence and absorption contrast samples. Has anybody out there experience with LED illumination in combination with a monochrome camera to generate 'true color' images? Your advice is welcome, commercial replies off list, please. Thank you, jens --- dr. jens rietdorf head microscopy facility novartis research foundation friedrich-miescher-institute, wro1066.2.16 maulbeerstr.66, 4058 basel, switzerland txt: rietdorf at fmi dot ch fon: +41.61.69.75172 fax: +41.61.69.73976 | 
|    | Martin Wessendorf-2 | 
|   | 
				In reply to this post by Rietdorf, Jens
			 
		Dear Jens--
 Rietdorf, Jens wrote: > we are looking for suitable illumination & camera/s for our Leica Z6 > MacroFluo. It is used for both fluorescence and absorption contrast > samples. > > Has anybody out there experience with LED illumination in combination > with a monochrome camera to generate 'true color' images? Clever idea! Sounds as if it'd work nicely for transmitted illumination and tinctorial staining. However, unless I'm misunderstanding you, I don't think it would work as well for fluorescence, since the emission wavelengths observed would be a function of the molecule rather than the excitation wavelength. Good luck-- Martin Wessendorf -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] | 
|    | Patrick Van Oostveldt | 
|   | 
				In reply to this post by Rietdorf, Jens
			 
		We got good results in microscopy with leds and RGB filters obtained  
 from Agfa Gevaert. These filters were used in earlier days to print and adapt colorphotos and put in lightpath of the magnifier to compensate color. They are named gelatine color filters and are availbale in different densities. Probably you can also use the color head from a photographic enlarger if available. You should test the exposure for different colors on a white paper and rescale the RGB images to obtain a white field. ImageJ allows this on the spot. Bye Patrick Van Oostveldt pictuesQuoting "Rietdorf, Jens" <[hidden email]>: > Dear colleagues, > > we are looking for suitable illumination & camera/s for our Leica Z6 > MacroFluo. It is used for both fluorescence and absorption contrast > samples. > > Has anybody out there experience with LED illumination in combination > with a monochrome camera to generate 'true color' images? > > Your advice is welcome, commercial replies off list, please. > > Thank you, jens > > --- > dr. jens rietdorf > head microscopy facility > novartis research foundation > friedrich-miescher-institute, wro1066.2.16 > maulbeerstr.66, 4058 basel, switzerland > txt: rietdorf at fmi dot ch > fon: +41.61.69.75172 > fax: +41.61.69.73976 > -- Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 | 
|    | Rietdorf, Jens | 
|   | 
				In reply to this post by Martin Wessendorf-2
			 
		Dear Martin and others,
 sorry, I see that I should be more clear. Indeed the question is if someone has experience in replacing a color camera by a combination of multi-LED illumination and a monochrome camera. We would prefer a monochrome camera, because this could then also be used for efficient fluorescence imaging and we would not have to switch between a color and a monochrome camera. For fluorescence we are using regular epi-illumination, for LED pseudo color we would use transmission illumination. Thanks, jens -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Martin Wessendorf Sent: Wednesday, October 28, 2009 3:46 PM To: [hidden email] Subject: Re: question: camera for macroscope Dear Jens-- Rietdorf, Jens wrote: > we are looking for suitable illumination & camera/s for our Leica Z6 > MacroFluo. It is used for both fluorescence and absorption contrast > samples. > > Has anybody out there experience with LED illumination in combination > with a monochrome camera to generate 'true color' images? Clever idea! Sounds as if it'd work nicely for transmitted illumination and tinctorial staining. However, unless I'm misunderstanding you, I don't think it would work as well for fluorescence, since the emission wavelengths observed would be a function of the molecule rather than the excitation wavelength. Good luck-- Martin Wessendorf -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] | 
|    | Peter Humphreys | 
|   | 
		We have an old optronics magnafire camera that has 3 colour filters
 (R,G,B) and blank for fluorescence. there are similar still on the market. With a white light LED the colour balance should only have to be set once. Peter Centre for Stem Cell Research Cambridge UK > Dear Martin and others, > > sorry, I see that I should be more clear. Indeed the question is if > someone has experience in replacing a color camera by a combination of > multi-LED illumination and a monochrome camera. > > We would prefer a monochrome camera, because this could then also be > used for efficient fluorescence imaging and we would not have to switch > between a color and a monochrome camera. > > For fluorescence we are using regular epi-illumination, for LED pseudo > color we would use transmission illumination. > > Thanks, jens > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Martin Wessendorf > Sent: Wednesday, October 28, 2009 3:46 PM > To: [hidden email] > Subject: Re: question: camera for macroscope > > Dear Jens-- > > Rietdorf, Jens wrote: > > > we are looking for suitable illumination & camera/s for our Leica Z6 > > MacroFluo. It is used for both fluorescence and absorption contrast > > samples. > > > > Has anybody out there experience with LED illumination in combination > > with a monochrome camera to generate 'true color' images? > > Clever idea! Sounds as if it'd work nicely for transmitted illumination > > and tinctorial staining. However, unless I'm misunderstanding you, I > don't think it would work as well for fluorescence, since the emission > wavelengths observed would be a function of the molecule rather than the > > excitation wavelength. > > Good luck-- > > Martin Wessendorf | 
|    | Jerry (Gerald) Sedgewick | 
|   | 
				In reply to this post by Rietdorf, Jens
			 
		I haven't exactly done this application, but it appears that there are 
 two ways to accomplish your end. Either you can use white light LEDs and then rotate red, green and blue filters in front of your camera, or you can choose combinations of LED lights to accomplish approximately the same end. I would advise the use of the former. R,G, B filters in front of cameras are fairly broad bandpass filters with some overlap (B=375nm to 525nm, G=475nm-600nm and R=600nm-675nm). The wavelengths that emit from colored light LEDs, on the other hand, are narrow gaussian profiles. While one could conceivably purchase a series of LEDs to fill out a wavelength range (e.g. for the blue wavelength range the following LEDs could be purchased: 385, 405, 455, 470 and 505nm), this would be expensive and still would not provide the broad illumination of R, G, B filter set with a white light source. Also, the white light LEDs do not emit at 500nm, but this is probably a moot point for capturing generic color images. Jerry Rietdorf, Jens wrote: > Dear Martin and others, > > sorry, I see that I should be more clear. Indeed the question is if > someone has experience in replacing a color camera by a combination of > multi-LED illumination and a monochrome camera. > > We would prefer a monochrome camera, because this could then also be > used for efficient fluorescence imaging and we would not have to switch > between a color and a monochrome camera. > > For fluorescence we are using regular epi-illumination, for LED pseudo > color we would use transmission illumination. > > Thanks, jens > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Martin Wessendorf > Sent: Wednesday, October 28, 2009 3:46 PM > To: [hidden email] > Subject: Re: question: camera for macroscope > > Dear Jens-- > > Rietdorf, Jens wrote: > > >> we are looking for suitable illumination & camera/s for our Leica Z6 >> MacroFluo. It is used for both fluorescence and absorption contrast >> samples. >> >> Has anybody out there experience with LED illumination in combination >> with a monochrome camera to generate 'true color' images? >> > > Clever idea! Sounds as if it'd work nicely for transmitted illumination > > and tinctorial staining. However, unless I'm misunderstanding you, I > don't think it would work as well for fluorescence, since the emission > wavelengths observed would be a function of the molecule rather than the > > excitation wavelength. > > Good luck-- > > Martin Wessendorf > -- Jerry (Gerald) Sedgewick Core Facility Director, Biomedical Image Processing Lab (BIPL) University of Minnesota, Department of Neuroscience 1-205 Hasselmo Hall 312 Church St. S.E. Minneapolis, MN 55455 612-624-6607 [hidden email] http://www.bipl.umn.edu Author: "Scientific Imaging with Photoshop: Methods, Measurement and Output." Rawlight.com (dba "Sedgewick Initiatives") 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 [hidden email] http://www.quickphotoshop.com http://www.rawlight.com http://www.jerrysedgewick.com | 
|    | Carol Heckman | 
|   | 
		There is a camera by QImaging that sounds ideal for this.  It is monochrome but uses a filter wheel with liquid crystal filters to get the bands in R, G, and B.  We purchased one and will know soon how it looks.  It looked good at the demo.
 Carol Heckman Center for Microscopy & Microanalysis Bowling Green State University ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Jerry (Gerald) Sedgewick [[hidden email]] Sent: Wednesday, October 28, 2009 2:15 PM To: [hidden email] Subject: Re: question: camera for macroscope I haven't exactly done this application, but it appears that there are two ways to accomplish your end. Either you can use white light LEDs and then rotate red, green and blue filters in front of your camera, or you can choose combinations of LED lights to accomplish approximately the same end. I would advise the use of the former. R,G, B filters in front of cameras are fairly broad bandpass filters with some overlap (B=375nm to 525nm, G=475nm-600nm and R=600nm-675nm). The wavelengths that emit from colored light LEDs, on the other hand, are narrow gaussian profiles. While one could conceivably purchase a series of LEDs to fill out a wavelength range (e.g. for the blue wavelength range the following LEDs could be purchased: 385, 405, 455, 470 and 505nm), this would be expensive and still would not provide the broad illumination of R, G, B filter set with a white light source. Also, the white light LEDs do not emit at 500nm, but this is probably a moot point for capturing generic color images. Jerry Rietdorf, Jens wrote: > Dear Martin and others, > > sorry, I see that I should be more clear. Indeed the question is if > someone has experience in replacing a color camera by a combination of > multi-LED illumination and a monochrome camera. > > We would prefer a monochrome camera, because this could then also be > used for efficient fluorescence imaging and we would not have to switch > between a color and a monochrome camera. > > For fluorescence we are using regular epi-illumination, for LED pseudo > color we would use transmission illumination. > > Thanks, jens > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Martin Wessendorf > Sent: Wednesday, October 28, 2009 3:46 PM > To: [hidden email] > Subject: Re: question: camera for macroscope > > Dear Jens-- > > Rietdorf, Jens wrote: > > >> we are looking for suitable illumination & camera/s for our Leica Z6 >> MacroFluo. It is used for both fluorescence and absorption contrast >> samples. >> >> Has anybody out there experience with LED illumination in combination >> with a monochrome camera to generate 'true color' images? >> > > Clever idea! Sounds as if it'd work nicely for transmitted illumination > > and tinctorial staining. However, unless I'm misunderstanding you, I > don't think it would work as well for fluorescence, since the emission > wavelengths observed would be a function of the molecule rather than the > > excitation wavelength. > > Good luck-- > > Martin Wessendorf > -- Jerry (Gerald) Sedgewick Core Facility Director, Biomedical Image Processing Lab (BIPL) University of Minnesota, Department of Neuroscience 1-205 Hasselmo Hall 312 Church St. S.E. Minneapolis, MN 55455 612-624-6607 [hidden email] http://www.bipl.umn.edu Author: "Scientific Imaging with Photoshop: Methods, Measurement and Output." Rawlight.com (dba "Sedgewick Initiatives") 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 [hidden email] http://www.quickphotoshop.com http://www.rawlight.com http://www.jerrysedgewick.com | 
|    | Jerry (Gerald) Sedgewick | 
|   | 
		Does anyone have experience with viewing necrotic tissues using DIC?  
 I'm wondering how tough it'll be to differentiate between live tissues and necrotic areas of tissue. I'm hoping to use a 0.70NA APO lens for DIC as well as some fluorescent images on the specimens without having to swap out the lenses, but I have to be able to pick out and avoid the necrotic areas of tissue. Thanks, Will Yutzy -- Will Yutzy [hidden email] <mailto:[hidden email]>, [hidden email] <mailto:[hidden email]> > > -- Jerry (Gerald) Sedgewick Core Facility Director, Biomedical Image Processing Lab (BIPL) University of Minnesota, Department of Neuroscience 1-205 Hasselmo Hall 312 Church St. S.E. Minneapolis, MN 55455 612-624-6607 [hidden email] http://www.bipl.umn.edu Author: "Scientific Imaging with Photoshop: Methods, Measurement and Output." Rawlight.com (dba "Sedgewick Initiatives") 965 Cromwell Avenue Saint Paul, MN 55114 651-788-2261 [hidden email] http://www.quickphotoshop.com http://www.rawlight.com http://www.jerrysedgewick.com | 