|
Ekaterina PAPUSHEVA-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Need your advice regarding the contamination of our MP cooling. We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. Are there any tips of how to get the growth out? We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. Does someone have experience with these two approaches, and is there something better? Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) Ekaterina ___ Ekaterina Papusheva, Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ekatrina I am not sure about ThermoTek brand chillers but we use a 10% alcohol and water in our recirculating chiller and have never had anything grow in it. Dan On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Need your advice regarding the contamination of our MP cooling. We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. Are there any tips of how to get the growth out? We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. Does someone have experience with these two approaches, and is there something better? Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) Ekaterina ___ Ekaterina Papusheva, Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. Dan Focht Bioptechs, Inc. 3560 Beck Rd. Butler, PA 16002 www.bioptechs.com P: (724)282-7145 F: (724)282-0745 [hidden email] |
|
Gary G. Li-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ekaterina, You need to add 10% OptiShield Plus (Opti Temp, Inc. www.optishield.net) in the distilled water to prevent biological growth in the coolant line. OptiShield Plus protects against both galvanic corrosion and biological buildup. OptiShield Plus, P/N 400113, 1 Pint Size; P/N 400115, 1 Gallon Size. Flush the coolant line more frequently until it becomes normal. Then make coolant change monthly afterwards. Gary G Li, PhD Microscopy Facility Manager NIH/NIDA at Baltimore MD 21224, USA On Fri, Oct 7, 2011 at 12:00 PM, Dan Focht <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ekatrina > > > I am not sure about ThermoTek brand chillers but we use a 10% alcohol and water in our recirculating chiller and have never had anything grow in it. > > > Dan > > > > > > On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Need your advice regarding the contamination of our MP cooling. > > We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. > Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. > > Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. > > Are there any tips of how to get the growth out? > We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. > > Does someone have experience with these two approaches, and is there something better? > > Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) > > Ekaterina > > ___ > > Ekaterina Papusheva, > Imaging Facility Manager > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. > > Dan Focht > Bioptechs, Inc. > 3560 Beck Rd. > Butler, PA 16002 > www.bioptechs.com > P: (724)282-7145 > F: (724)282-0745 > [hidden email] > |
|
Oshel, Philip Eugene |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ekaterina, What is the plumbing for the chiller lines? Opaque hose or transparent? Polymer hose or metal pipes? If transparent polymer, then maybe you do have algae or bacteria, possibly with opaque hose. Check with the chiller manufacturer, but I have had success with biocontamination in chiller lines by running about 1/2 a filter-holder's worth of bleach through the lines, then rinse with water. BUT! check first - this can be a horrible idea. If there are metal pipes anywhere in the system, and 2 different metals are in contact, like copper lines connecting to the coils of the chiller, even if polymer hoses are used otherwise, then you could be getting electrolytic corrosion. This can occur even if you don't see any obvious metal connections (what's inside the chiller?). The simplest thing is to put some of the green gunk under a microscope and look. If you have access to an SEM with EDS (x-ray spectroscopy), you can get a quick analysis of the stuff. Also, check the distilled water and the manufacturer's requirements. Distilled water is usually acidic (from atmospheric CO2 uptake) which can cause problems, and is also know as "hungry water" (as is deionized water). The lack of ions in the water, especially metals, causes the water to "pull" metal ions out of the pipes to get some chemical equilibrium. This isn't a problem for systems specifically designed for distilled water, but can be for systems that aren't so designed. Phil >Dear all, > >Need your advice regarding the contamination of our MP cooling. > >We were filling our ThermoTek chillers (T255P) that are connected >with Cameleon II MP lasers with distilled water, according to >Coherent instructions. >Since few days, we got dramatic contamination (some algae or >bacteria growing) in the water. > >Even if we flash a chiller with 10 l water and refill new, we get >already some visible contamination next day. One of the systems was >clogged once. > >Are there any tips of how to get the growth out? >We thought of adding 5% ethanol instead of water. Or building >bacterial filters (like on FACS machines) in the flow circle. > >Does someone have experience with these two approaches, and is there >something better? > >Looking forward to many helpful tips! Nice weekend to everyone, >except those who have to flush their chillers :) > >Ekaterina > >___ > >Ekaterina Papusheva, >Imaging Facility Manager > >Institute of Science and Technology Austria -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
|
Armstrong, Brian |
|
|
In reply to this post by Gary G. Li-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, I concur with Gary. We use Optishield and change the distilled water monthly. Note; this is not milliQ or DDH2O. I wish that our first Coherent service rep would have told me this. We had similar issues with clogging etc... and had to replace the cooler (which Coherent will do free of charge via service contract). Since I have been changing the water monthly it has not been a problem. Cheers, Brian Armstrong PhD Light Microscopy Core Beckman Research Institute 1450 East Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/SharedResources/LightMicroscopy/LightMicroHome.htm -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Gary G. Li Sent: Friday, October 07, 2011 9:03 AM To: [hidden email] Subject: Re: recurring contamination of MP cooling ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ekaterina, You need to add 10% OptiShield Plus (Opti Temp, Inc. www.optishield.net) in the distilled water to prevent biological growth in the coolant line. OptiShield Plus protects against both galvanic corrosion and biological buildup. OptiShield Plus, P/N 400113, 1 Pint Size; P/N 400115, 1 Gallon Size. Flush the coolant line more frequently until it becomes normal. Then make coolant change monthly afterwards. Gary G Li, PhD Microscopy Facility Manager NIH/NIDA at Baltimore MD 21224, USA On Fri, Oct 7, 2011 at 12:00 PM, Dan Focht <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ekatrina > > > I am not sure about ThermoTek brand chillers but we use a 10% alcohol and water in our recirculating chiller and have never had anything grow in it. > > > Dan > > > > > > On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Need your advice regarding the contamination of our MP cooling. > > We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. > Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. > > Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. > > Are there any tips of how to get the growth out? > We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. > > Does someone have experience with these two approaches, and is there something better? > > Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) > > Ekaterina > > ___ > > Ekaterina Papusheva, > Imaging Facility Manager > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. > > Dan Focht > Bioptechs, Inc. > 3560 Beck Rd. > Butler, PA 16002 > www.bioptechs.com > P: (724)282-7145 > F: (724)282-0745 > [hidden email] > --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
|
Patrick Van Oostveldt |
|
|
In reply to this post by Dan Focht
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Including some part of the tubing by a copperpipe can solve a lot. If more money is available a small silver pipe can also be used. Patrick Quoting "Dan Focht" <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ekatrina > > > I am not sure about ThermoTek brand chillers but we use a 10% > alcohol and water in our recirculating chiller and have never had > anything grow in it. > > > Dan > > > > > > On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Need your advice regarding the contamination of our MP cooling. > > We were filling our ThermoTek chillers (T255P) that are connected > with Cameleon II MP lasers with distilled water, according to > Coherent instructions. > Since few days, we got dramatic contamination (some algae or > bacteria growing) in the water. > > Even if we flash a chiller with 10 l water and refill new, we get > already some visible contamination next day. One of the systems was > clogged once. > > Are there any tips of how to get the growth out? > We thought of adding 5% ethanol instead of water. Or building > bacterial filters (like on FACS machines) in the flow circle. > > Does someone have experience with these two approaches, and is there > something better? > > Looking forward to many helpful tips! Nice weekend to everyone, > except those who have to flush their chillers :) > > Ekaterina > > ___ > > Ekaterina Papusheva, > Imaging Facility Manager > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are > confidential and intended only for the individual named. If you have > received the Message in error you should not distribute or copy this > Message. Please notify the sender immediately by e-mail if you have > received this Message in error and delete the Message from your > system. Correspondence via e-mail is for information purposes. IST > Austria neither makes nor accepts legally binding statements unless > explicitly otherwise agreed to the contrary. > > Dan Focht > Bioptechs, Inc. > 3560 Beck Rd. > Butler, PA 16002 > www.bioptechs.com > P: (724)282-7145 > F: (724)282-0745 > [hidden email] > -- Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Panacide is what I've used. It came from BDH, and I imagine it's still available. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: Saturday, 8 October 2011 2:06 AM To: [hidden email] Subject: recurring contamination of MP cooling ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Need your advice regarding the contamination of our MP cooling. We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. Are there any tips of how to get the growth out? We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. Does someone have experience with these two approaches, and is there something better? Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) Ekaterina ___ Ekaterina Papusheva, Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1410 / Virus Database: 1520/3942 - Release Date: 10/06/11 |
|
Mark Cannell |
|
|
In reply to this post by Patrick Van Oostveldt
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Do you have a reference for the antifungal activity of simple metallic silver? I've heard this repeated many times but never found a reputable source. Cheers On 7/10/2011, at 10:26 PM, Patrick Van Oostveldt wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Including some part of the tubing by a copperpipe can solve a lot. > If more money is available a small silver pipe can also be used. > > Patrick > > Quoting "Dan Focht" <[hidden email]>: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Ekatrina >> >> >> I am not sure about ThermoTek brand chillers but we use a 10% alcohol and water in our recirculating chiller and have never had anything grow in it. >> >> >> Dan >> >> >> >> >> >> On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> Need your advice regarding the contamination of our MP cooling. >> >> We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. >> Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. >> >> Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. >> >> Are there any tips of how to get the growth out? >> We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. >> >> Does someone have experience with these two approaches, and is there something better? >> >> Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) >> >> Ekaterina >> >> ___ >> >> Ekaterina Papusheva, >> Imaging Facility Manager >> >> Institute of Science and Technology Austria >> >> Phone +43-(0)2243 9000-1043; >> Mobile +43-(0)664 88509130 >> Fax +43-(0)2243 9000-2000 >> >> Am Campus 1 >> >> A-3400 Klosterneuburg >> >> Visit our website: www.ist.ac.at<http://www.ist.ac.at> >> >> >> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. >> >> Dan Focht >> Bioptechs, Inc. >> 3560 Beck Rd. >> Butler, PA 16002 >> www.bioptechs.com >> P: (724)282-7145 >> F: (724)282-0745 >> [hidden email] >> > > > > -- > Dep. Moleculaire Biotechnologie > Coupure links 653 > B 9000 GENT > > tel 09 264 5969 > fax 09 264 6219 |
|
David Baddeley |
|
|
In reply to this post by Patrick Van Oostveldt
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** you might want to run it past the manufacturer first, but we've used antifreeze to good effect in a closed circuit camera cooling loop. cheers, David ________________________________ From: Patrick Van Oostveldt <[hidden email]> To: [hidden email] Sent: Saturday, 8 October 2011 10:26 AM Subject: Re: recurring contamination of MP cooling ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Including some part of the tubing by a copperpipe can solve a lot. If more money is available a small silver pipe can also be used. Patrick Quoting "Dan Focht" <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Ekatrina > > > I am not sure about ThermoTek brand chillers but we use a 10% alcohol and water in our recirculating chiller and have never had anything grow in it. > > > Dan > > > > > > On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Need your advice regarding the contamination of our MP cooling. > > We were filling our ThermoTek chillers (T255P) that are connected with Cameleon II MP lasers with distilled water, according to Coherent instructions. > Since few days, we got dramatic contamination (some algae or bacteria growing) in the water. > > Even if we flash a chiller with 10 l water and refill new, we get already some visible contamination next day. One of the systems was clogged once. > > Are there any tips of how to get the growth out? > We thought of adding 5% ethanol instead of water. Or building bacterial filters (like on FACS machines) in the flow circle. > > Does someone have experience with these two approaches, and is there something better? > > Looking forward to many helpful tips! Nice weekend to everyone, except those who have to flush their chillers :) > > Ekaterina > > ___ > > Ekaterina Papusheva, > Imaging Facility Manager > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. > > Dan Focht > Bioptechs, Inc. > 3560 Beck Rd. > Butler, PA 16002 > www.bioptechs.com > P: (724)282-7145 > F: (724)282-0745 > [hidden email] > --Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 |
|
Patrick Van Oostveldt |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Here some references at least for microbes. For furtehr details look at the website of our colleague Willy Verstraete. If yoy like specific antifungal effects, this is more complex. But what is routinely used is treatment of the containers with ultrasound, which inhibits also development of algae. References: Sintubin, L., B. De Gusseme, et al. (2011). "The antibacterial activity of biogenic silver and its mode of action." Applied Microbiology and Biotechnology 91(1): 153-162. Chopra, I (2007). "The increasing use of silver-based products as antimicrobial agents: a useful development or a cause for concern?". The Journal of antimicrobial chemotherapy 59 (4): 587?90. Greetings, Patrick Van Oostveldt Quoting "Mark Cannell" <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Do you have a reference for the antifungal activity of simple > metallic silver? I've heard this repeated many times but never found > a reputable source. > > Cheers > > > On 7/10/2011, at 10:26 PM, Patrick Van Oostveldt wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Including some part of the tubing by a copperpipe can solve a lot. >> If more money is available a small silver pipe can also be used. >> >> Patrick >> >> Quoting "Dan Focht" <[hidden email]>: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Ekatrina >>> >>> >>> I am not sure about ThermoTek brand chillers but we use a 10% >>> alcohol and water in our recirculating chiller and have never had >>> anything grow in it. >>> >>> >>> Dan >>> >>> >>> >>> >>> >>> On Oct 7, 2011, at 11:05 AM, Ekaterina PAPUSHEVA wrote: >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear all, >>> >>> Need your advice regarding the contamination of our MP cooling. >>> >>> We were filling our ThermoTek chillers (T255P) that are connected >>> with Cameleon II MP lasers with distilled water, according to >>> Coherent instructions. >>> Since few days, we got dramatic contamination (some algae or >>> bacteria growing) in the water. >>> >>> Even if we flash a chiller with 10 l water and refill new, we get >>> already some visible contamination next day. One of the systems >>> was clogged once. >>> >>> Are there any tips of how to get the growth out? >>> We thought of adding 5% ethanol instead of water. Or building >>> bacterial filters (like on FACS machines) in the flow circle. >>> >>> Does someone have experience with these two approaches, and is >>> there something better? >>> >>> Looking forward to many helpful tips! Nice weekend to everyone, >>> except those who have to flush their chillers :) >>> >>> Ekaterina >>> >>> ___ >>> >>> Ekaterina Papusheva, >>> Imaging Facility Manager >>> >>> Institute of Science and Technology Austria >>> >>> Phone +43-(0)2243 9000-1043; >>> Mobile +43-(0)664 88509130 >>> Fax +43-(0)2243 9000-2000 >>> >>> Am Campus 1 >>> >>> A-3400 Klosterneuburg >>> >>> Visit our website: www.ist.ac.at<http://www.ist.ac.at> >>> >>> >>> This message and any attachment (together "the Message") are >>> confidential and intended only for the individual named. If you >>> have received the Message in error you should not distribute or >>> copy this Message. Please notify the sender immediately by e-mail >>> if you have received this Message in error and delete the Message >>> from your system. Correspondence via e-mail is for information >>> purposes. IST Austria neither makes nor accepts legally binding >>> statements unless explicitly otherwise agreed to the contrary. >>> >>> Dan Focht >>> Bioptechs, Inc. >>> 3560 Beck Rd. >>> Butler, PA 16002 >>> www.bioptechs.com >>> P: (724)282-7145 >>> F: (724)282-0745 >>> [hidden email] >>> >> >> >> >> -- >> Dep. Moleculaire Biotechnologie >> Coupure links 653 >> B 9000 GENT >> >> tel 09 264 5969 >> fax 09 264 6219 > -- Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 |
| Free forum by Nabble | Edit this page |