http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear All, I am trying to collect some Z sections on a thick unstained sample from spinal disc using reflective mode. I would appreciate any input from your own experiences with reflection contrast. The disc material is mainly made of very dense connective tissues and some cross-ties collagen fibres. I am mainly interested to see the detail of collagen fibres between the intervertebral disc lamellae. Obviously the sample is not immunofluorescently labelled. Do you think normal H&E staining would help in reflective mode of confocal imaging? What laser line would work better, Ar 488, or green HeNe 546? With regards, Ghafar *********************************************** Ghafar T Sarvestani, PhD Hanson Institute for Medical Research Hanson Building level 2, Room 2-42-N IMVS, Frome Road, Adelaide SA 5000 Tel: +61 8 82223603 Fax: +61 8 82223162 Email:[hidden email] This e-mail may contain confidential information, which also may be legally privileged. Only the intended recipient(s) may access, use, distribute or copy this e-mail. If this e-mail is received in error, please inform the sender by return e-mail and delete the original. It is the recipient's responsibility to check the e-mail and any attached files for viruses. |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Without having practical experience I would assume: Scattering: Smaller particles / longer wavelengths scatter more uniformly and since you would want the opposite (more forward and backward scattering) you might want to go with shorter wavelengths. However it's likely that the gain is minimal as the ratio between fiber diameter and wavelength does not change much from Ar to HeNe and you might not get more light back. Absorption: What is more critical is absorption and the more you have, the more attenuation (less penetration) you will have. In that sense H&E staining would not be useful and a red HeNe or a laser between 700 and 900 nms would be best. Perhaps a silver or gold stain could help for reflection. I hope someone with practical experience will respond. Urs -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ghafar Sarvestani Sent: Wednesday, January 09, 2008 7:59 PM To: [hidden email] Subject: reflection contrast Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear All, I am trying to collect some Z sections on a thick unstained sample from spinal disc using reflective mode. I would appreciate any input from your own experiences with reflection contrast. The disc material is mainly made of very dense connective tissues and some cross-ties collagen fibres. I am mainly interested to see the detail of collagen fibres between the intervertebral disc lamellae. Obviously the sample is not immunofluorescently labelled. Do you think normal H&E staining would help in reflective mode of confocal imaging? What laser line would work better, Ar 488, or green HeNe 546? With regards, Ghafar *********************************************** Ghafar T Sarvestani, PhD Hanson Institute for Medical Research Hanson Building level 2, Room 2-42-N IMVS, Frome Road, Adelaide SA 5000 Tel: +61 8 82223603 Fax: +61 8 82223162 Email:[hidden email] This e-mail may contain confidential information, which also may be legally privileged. Only the intended recipient(s) may access, use, distribute or copy this e-mail. If this e-mail is received in error, please inform the sender by return e-mail and delete the original. It is the recipient's responsibility to check the e-mail and any attached files for viruses. |
|
|
In reply to this post by Ghafar Sarvestani
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Ghafar, I'd try second harmonic if I were you. It works best in transmission but lots of people have used it in back-propagated mode. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ghafar Sarvestani Sent: Thursday, 10 January 2008 1:59 PM To: [hidden email] Subject: reflection contrast Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear All, I am trying to collect some Z sections on a thick unstained sample from spinal disc using reflective mode. I would appreciate any input from your own experiences with reflection contrast. The disc material is mainly made of very dense connective tissues and some cross-ties collagen fibres. I am mainly interested to see the detail of collagen fibres between the intervertebral disc lamellae. Obviously the sample is not immunofluorescently labelled. Do you think normal H&E staining would help in reflective mode of confocal imaging? What laser line would work better, Ar 488, or green HeNe 546? With regards, Ghafar *********************************************** Ghafar T Sarvestani, PhD Hanson Institute for Medical Research Hanson Building level 2, Room 2-42-N IMVS, Frome Road, Adelaide SA 5000 Tel: +61 8 82223603 Fax: +61 8 82223162 Email:[hidden email] This e-mail may contain confidential information, which also may be legally privileged. Only the intended recipient(s) may access, use, distribute or copy this e-mail. If this e-mail is received in error, please inform the sender by return e-mail and delete the original. It is the recipient's responsibility to check the e-mail and any attached files for viruses. No virus found in this incoming message. Checked by AVG Free Edition. Version: 7.5.516 / Virus Database: 269.19.0/1216 - Release Date: 9/01/2008 10:16 AM No virus found in this outgoing message. Checked by AVG Free Edition. Version: 7.5.516 / Virus Database: 269.19.0/1216 - Release Date: 9/01/2008 10:16 AM |
|
|
In reply to this post by Ghafar Sarvestani
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Ghafar, If you have an empty position on the emission side, try each laser line in turn. Collagen fibrils can show up looking like needles in live cells 3D matrix experiments (see also J. Paul Robinson's papers). Different laser lines will have different artifacts - easiest to see these by focusing on the coverglass - medium interface. George At 09:58 PM 1/9/2008, you wrote: >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Dear All, >I am trying to collect some Z sections on a thick unstained sample from >spinal disc using reflective mode. I would appreciate any input from your >own experiences with reflection contrast. The disc material is mainly made >of very dense connective tissues and some cross-ties collagen fibres. I am >mainly interested to see the detail of collagen fibres between the >intervertebral disc lamellae. Obviously the sample is not >immunofluorescently labelled. Do you think normal H&E staining would help in >reflective mode of confocal imaging? What laser line would work better, Ar >488, or green HeNe 546? >With regards, >Ghafar > >*********************************************** >Ghafar T Sarvestani, PhD >Hanson Institute for Medical Research >Hanson Building level 2, Room 2-42-N >IMVS, Frome Road, Adelaide SA 5000 > >Tel: +61 8 82223603 >Fax: +61 8 82223162 >Email:[hidden email] > >This e-mail may contain confidential information, which also >may be legally privileged. Only the intended recipient(s) may access, use, >distribute or copy this e-mail. If this e-mail is received in error, please >inform the sender by return e-mail and delete the original. It is the >recipient's responsibility to check the e-mail and any attached files for >viruses. George McNamara, Ph.D. University of Miami, Miller School of Medicine Image Core Miami, FL 33010 [hidden email] [hidden email] 305-243-8436 office http://home.earthlink.net/~pubspectra/ http://home.earthlink.net/~geomcnamara/ http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility) |
|
|
In reply to this post by Ghafar Sarvestani
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Ghafar, We did quite a bit of reflected confocal on a homemade system at Northeastern. We put a quarter wave plate in the path to reduce back reflections from optics/glass. In terms of wavelength, a good deal depends on the coatings of your optics. We got our best images around 750nm. Second harmomic also produced some pretty nice images of the collagen. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ghafar Sarvestani Sent: Wednesday, January 09, 2008 9:59 PM To: [hidden email] Subject: reflection contrast Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear All, I am trying to collect some Z sections on a thick unstained sample from spinal disc using reflective mode. I would appreciate any input from your own experiences with reflection contrast. The disc material is mainly made of very dense connective tissues and some cross-ties collagen fibres. I am mainly interested to see the detail of collagen fibres between the intervertebral disc lamellae. Obviously the sample is not immunofluorescently labelled. Do you think normal H&E staining would help in reflective mode of confocal imaging? What laser line would work better, Ar 488, or green HeNe 546? With regards, Ghafar *********************************************** Ghafar T Sarvestani, PhD Hanson Institute for Medical Research Hanson Building level 2, Room 2-42-N IMVS, Frome Road, Adelaide SA 5000 Tel: +61 8 82223603 Fax: +61 8 82223162 Email:[hidden email] This e-mail may contain confidential information, which also may be legally privileged. Only the intended recipient(s) may access, use, distribute or copy this e-mail. If this e-mail is received in error, please inform the sender by return e-mail and delete the original. It is the recipient's responsibility to check the e-mail and any attached files for viruses. |