Hello everybody

We are looking to buy a refractometer for our lab, the goal is to measure TDE refractive index for clearing and mounting samples.  Can anybody recommend a good one from experience?  There's many choices available it seems.

Thanks very much in advance,

Yevgeniy Romin

=====================================================================

     Please note that this e-mail and any files transmitted from
     Memorial Sloan Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.
Mark A. Sanders Mark A. Sanders
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Greetings,

I have been happy with this one for measuring mounting media.
https://www.fishersci.com/shop/products/atago-digital-hand-held-pocket-refractometer-pal-ri/05100187 <https://www.fishersci.com/shop/products/atago-digital-hand-held-pocket-refractometer-pal-ri/05100187>

Cheers,
Mark
************************************************
Mark A. Sanders      University of Minnesota
Program Director      Twin Cities Campus
University Imaging Centers  
Nikon Center of Excellence <https://www.nikoninstruments.com/Imaging-Centers/Nikon-Centers-of-Excellence/Nikon-CofE-Locations/Americas/University-of-Minnesota>
Thermo Fisher Scientific Supply Center <http://uic.umn.edu/content/thermofisher-scientific-supply-center>
www.uic.umn.edu
http://uic.umn.edu/content/locations
Acknowledgement in your publications and presentations of work performed in the UIC plays a vital role in securing support and the funding necessary to maintain and operate this valuable research resource.   For publications that were made possible by work performed in the UIC, please use the acknowledgement statement ‘We thank the University of Minnesota, University Imaging Centers for experimental and technical support.’

Please also consider UIC staff for co-authorship if they played a key role in the study.  
Thank you.

> On Jan 30, 2018, at 2:31 PM, [hidden email] wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hello everybody
>
> We are looking to buy a refractometer for our lab, the goal is to measure TDE refractive index for clearing and mounting samples.  Can anybody recommend a good one from experience?  There's many choices available it seems.
>
> Thanks very much in advance,
>
> Yevgeniy Romin
>
> =====================================================================
>
>     Please note that this e-mail and any files transmitted from
>     Memorial Sloan Kettering Cancer Center may be privileged, confidential,
>     and protected from disclosure under applicable law. If the reader of
>     this message is not the intended recipient, or an employee or agent
>     responsible for delivering this message to the intended recipient,
>     you are hereby notified that any reading, dissemination, distribution,
>     copying, or other use of this communication or any of its attachments
>     is strictly prohibited.  If you have received this communication in
>     error, please notify the sender immediately by replying to this message
>     and deleting this message, any attachments, and all copies and backups
>     from your computer.
Zdenek Svindrych-2 Zdenek Svindrych-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

In reply to this post by Romin, Yevgeniy/Sloan Kettering Institute
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Yevgeniy,
I got one just like this (but perfectly working):

https://www.ebay.com/itm/Fisher-Scientific-Abbe-Benchtop-Refractometer-Parts
-or-Repair/292401156681


and was happy with it. Easy to clean, the RI range is from less than 1.3 to
more than 1.6. For $50 it was a good deal :-).

Best, zdenek

--
Zdenek Svindrych, Ph.D.
Research Associate - Imaging Specialist
Department of Biochemistry and Cell Biology
Geisel School of Medicine at Dartmouth
email: [hidden email]

---------- Původní e-mail ----------
Od: [hidden email]
Komu: [hidden email]
Datum: 30. 1. 2018 15:32:09
Předmět: refractometer question
"*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello everybody

We are looking to buy a refractometer for our lab, the goal is to measure
TDE refractive index for clearing and mounting samples. Can anybody
recommend a good one from experience? There's many choices available it
seems.

Thanks very much in advance,

Yevgeniy Romin

=====================================================================

Please note that this e-mail and any files transmitted from
Memorial Sloan Kettering Cancer Center may be privileged, confidential,
and protected from disclosure under applicable law. If the reader of
this message is not the intended recipient, or an employee or agent
responsible for delivering this message to the intended recipient,
you are hereby notified that any reading, dissemination, distribution,
copying, or other use of this communication or any of its attachments
is strictly prohibited. If you have received this communication in
error, please notify the sender immediately by replying to this message
and deleting this message, any attachments, and all copies and backups
from your computer.
"
Romain Laine Romain Laine
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi,
I can recommend the Bellingham & Stanley Abbe 5 refractometer
https://uk.vwr.com/store/product/2058765/abbe-refractometer

Cheers,

Romain


On Tue, Jan 30, 2018 at 10:13 PM, <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Yevgeniy,
> I got one just like this (but perfectly working):
>
> https://www.ebay.com/itm/Fisher-Scientific-Abbe-
> Benchtop-Refractometer-Parts
> -or-Repair/292401156681
>
>
> and was happy with it. Easy to clean, the RI range is from less than 1.3 to
> more than 1.6. For $50 it was a good deal :-).
>
> Best, zdenek
>
> --
> Zdenek Svindrych, Ph.D.
> Research Associate - Imaging Specialist
> Department of Biochemistry and Cell Biology
> Geisel School of Medicine at Dartmouth
> email: [hidden email]
>
> ---------- Původní e-mail ----------
> Od: [hidden email]
> Komu: [hidden email]
> Datum: 30. 1. 2018 15:32:09
> Předmět: refractometer question
> "*****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hello everybody
>
> We are looking to buy a refractometer for our lab, the goal is to measure
> TDE refractive index for clearing and mounting samples. Can anybody
> recommend a good one from experience? There's many choices available it
> seems.
>
> Thanks very much in advance,
>
> Yevgeniy Romin
>
> =====================================================================
>
> Please note that this e-mail and any files transmitted from
> Memorial Sloan Kettering Cancer Center may be privileged, confidential,
> and protected from disclosure under applicable law. If the reader of
> this message is not the intended recipient, or an employee or agent
> responsible for delivering this message to the intended recipient,
> you are hereby notified that any reading, dissemination, distribution,
> copying, or other use of this communication or any of its attachments
> is strictly prohibited. If you have received this communication in
> error, please notify the sender immediately by replying to this message
> and deleting this message, any attachments, and all copies and backups
> from your computer.
> "
>


<https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
Virus-free.
www.avast.com
<https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
<#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
Romin, Yevgeniy/Sloan Kettering Institute Romin, Yevgeniy/Sloan Kettering Institute
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thank you everybody for your recommendations, it was very helpful

Yevgeniy

> On Jan 31, 2018, at 6:27 AM, Romain Laine <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi,
> I can recommend the Bellingham & Stanley Abbe 5 refractometer
> https://uk.vwr.com/store/product/2058765/abbe-refractometer
>
> Cheers,
>
> Romain
>
>
>> On Tue, Jan 30, 2018 at 10:13 PM, <[hidden email]> wrote:
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hi Yevgeniy,
>> I got one just like this (but perfectly working):
>>
>> https://www.ebay.com/itm/Fisher-Scientific-Abbe-
>> Benchtop-Refractometer-Parts
>> -or-Repair/292401156681
>>
>>
>> and was happy with it. Easy to clean, the RI range is from less than 1.3 to
>> more than 1.6. For $50 it was a good deal :-).
>>
>> Best, zdenek
>>
>> --
>> Zdenek Svindrych, Ph.D.
>> Research Associate - Imaging Specialist
>> Department of Biochemistry and Cell Biology
>> Geisel School of Medicine at Dartmouth
>> email: [hidden email]
>>
>> ---------- Původní e-mail ----------
>> Od: [hidden email]
>> Komu: [hidden email]
>> Datum: 30. 1. 2018 15:32:09
>> Předmět: refractometer question
>> "*****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Hello everybody
>>
>> We are looking to buy a refractometer for our lab, the goal is to measure
>> TDE refractive index for clearing and mounting samples. Can anybody
>> recommend a good one from experience? There's many choices available it
>> seems.
>>
>> Thanks very much in advance,
>>
>> Yevgeniy Romin
>>
>> =====================================================================
>>
>> Please note that this e-mail and any files transmitted from
>> Memorial Sloan Kettering Cancer Center may be privileged, confidential,
>> and protected from disclosure under applicable law. If the reader of
>> this message is not the intended recipient, or an employee or agent
>> responsible for delivering this message to the intended recipient,
>> you are hereby notified that any reading, dissemination, distribution,
>> copying, or other use of this communication or any of its attachments
>> is strictly prohibited. If you have received this communication in
>> error, please notify the sender immediately by replying to this message
>> and deleting this message, any attachments, and all copies and backups
>> from your computer.
>> "
>>
>
>
> <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
> Virus-free.
> www.avast.com
> <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
> <#DAB4FAD8-2DD7-40BB-A1B8-4E2AA1F9FDF2>
>
>
>
> *** Only open attachments or links from trusted senders. Report phishing to [hidden email] ***

=====================================================================

     Please note that this e-mail and any files transmitted from
     Memorial Sloan Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.
Steffen Dietzel Steffen Dietzel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

In reply to this post by Romin, Yevgeniy/Sloan Kettering Institute
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We got the Abbe-refractometer ORT 1RS  from Kern:

https://www.carlroth.com/de/en/Current-Mailing/Special-Prices/Labware/Optical-Instruments/Abbe-refractometer-ORT-1RS/Abbe-refractometer-ORT-1RS/p/00000011000149eb00050023_en

couldn't resist when they had a special offer recently. So far we are
happy with it.

Steffen


Am 30.01.2018 um 21:31 schrieb [hidden email]:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hello everybody
>
> We are looking to buy a refractometer for our lab, the goal is to measure TDE refractive index for clearing and mounting samples.  Can anybody recommend a good one from experience?  There's many choices available it seems.
>
> Thanks very much in advance,
>
> Yevgeniy Romin
>
> =====================================================================
>
>       Please note that this e-mail and any files transmitted from
>       Memorial Sloan Kettering Cancer Center may be privileged, confidential,
>       and protected from disclosure under applicable law. If the reader of
>       this message is not the intended recipient, or an employee or agent
>       responsible for delivering this message to the intended recipient,
>       you are hereby notified that any reading, dissemination, distribution,
>       copying, or other use of this communication or any of its attachments
>       is strictly prohibited.  If you have received this communication in
>       error, please notify the sender immediately by replying to this message
>       and deleting this message, any attachments, and all copies and backups
>       from your computer.
>

--
-- ----------------------------------------------------------

Steffen Dietzel, PD Dr. rer. nat.
Head of the Core Facility Bioimaging at the Biomedical Center
Ludwig-Maximilians-Universität München
Walter-Brendel-Zentrum für Experimentelle Medizin

Address:
Biomedical Center
Großhaderner Straße 9
D-82152 Planegg-Martinsried

Phone: +49/89/2180-71518
skype: steffendietzel
e-mail: [hidden email]
fax-to-e-mail: +49/89/2180-9971518
http://www.bioimaging.bmc.med.uni-muenchen.de


--
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Biomedical Center (BMC)
Head of the Core Facility Bioimaging

Großhaderner Straße 9
D-82152 Planegg-Martinsried
Germany

http://www.bioimaging.bmc.med.uni-muenchen.de
Romin, Yevgeniy/Sloan Kettering Institute Romin, Yevgeniy/Sloan Kettering Institute
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: refractometer question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Great, thank you, this one does look nice, and the RI range is more then decent
________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of Steffen Dietzel [[hidden email]]
Sent: Wednesday, January 31, 2018 10:32 AM
To: [hidden email]
Subject: [EXTERNAL] Re: refractometer question

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We got the Abbe-refractometer ORT 1RS  from Kern:

https://www.carlroth.com/de/en/Current-Mailing/Special-Prices/Labware/Optical-Instruments/Abbe-refractometer-ORT-1RS/Abbe-refractometer-ORT-1RS/p/00000011000149eb00050023_en

couldn't resist when they had a special offer recently. So far we are
happy with it.

Steffen


Am 30.01.2018 um 21:31 schrieb [hidden email]:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hello everybody
>
> We are looking to buy a refractometer for our lab, the goal is to measure TDE refractive index for clearing and mounting samples.  Can anybody recommend a good one from experience?  There's many choices available it seems.
>
> Thanks very much in advance,
>
> Yevgeniy Romin
>
> =====================================================================
>
>       Please note that this e-mail and any files transmitted from
>       Memorial Sloan Kettering Cancer Center may be privileged, confidential,
>       and protected from disclosure under applicable law. If the reader of
>       this message is not the intended recipient, or an employee or agent
>       responsible for delivering this message to the intended recipient,
>       you are hereby notified that any reading, dissemination, distribution,
>       copying, or other use of this communication or any of its attachments
>       is strictly prohibited.  If you have received this communication in
>       error, please notify the sender immediately by replying to this message
>       and deleting this message, any attachments, and all copies and backups
>       from your computer.
>

--
-- ----------------------------------------------------------

Steffen Dietzel, PD Dr. rer. nat.
Head of the Core Facility Bioimaging at the Biomedical Center
Ludwig-Maximilians-Universität München
Walter-Brendel-Zentrum für Experimentelle Medizin

Address:
Biomedical Center
Großhaderner Straße 9
D-82152 Planegg-Martinsried

Phone: +49/89/2180-71518
skype: steffendietzel
e-mail: [hidden email]
fax-to-e-mail: +49/89/2180-9971518
http://www.bioimaging.bmc.med.uni-muenchen.de


--
------------------------------------------------------------
Steffen Dietzel, PD Dr. rer. nat
Ludwig-Maximilians-Universität München
Biomedical Center (BMC)
Head of the Core Facility Bioimaging

Großhaderner Straße 9
D-82152 Planegg-Martinsried
Germany

http://www.bioimaging.bmc.med.uni-muenchen.de



*** Only open attachments or links from trusted senders. Report phishing to [hidden email] ***

=====================================================================

     Please note that this e-mail and any files transmitted from
     Memorial Sloan Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.