Jurkevic, Aleksandr |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello All, We are about to renovate the room for two-photon microscopy and I wonder what suggestions you might have about reducing light pollution that affects NDDs. As one of the measures, we are planning to paint walls and ceiling in black. What about "small" things like aluminum electrical channels, wall shelves, air vents? Have they also to be painted in dark colors or that would be an overkill? I would appreciate if you share your solutions. Alexander Alexander Jurkevic, PhD Associate Director Molecular Cytology Core University of Missouri 120 Life Sciences Center 1201 E. Rollins St. Columbia, MO 65211-7310 Phone: 573-882-4895 Fax: 573-884-9676 |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** My own experience with this very problem is that in terms of the room, it is mainly the walls that matter. I have shiny aluminum shelves over the `scope, and light grey vents and the like in the ceiling, and it doesn't seem to make much difference. One key detail is that you want the walls painted in matte black, not glossy. The contractors might mix this up so be very specific. Now our biggest issue is the clothing worn by the researchers; they were wearing white lab coats which was scattering light from the computer monitors back towards the instruments. I have dark navy blue lab coats on hand that solve that problem, or the researchers turn the monitor off/away for long runs. If what you are doing is very picky (low signal, lots of gain) a little bit of blackout cloth suspended in front of the stage can help as well. We've (carefully) draped a bit over the eyepieces on occasion, letting it hang down in front of the stage as a quick but effective light shield. Craig On Mon, Dec 9, 2013 at 5:40 PM, Jurkevic, Aleksandr <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All, > > We are about to renovate the room for two-photon microscopy and I wonder > what suggestions you might have about reducing light pollution that affects > NDDs. As one of the measures, we are planning to paint walls and ceiling in > black. What about "small" things like aluminum electrical channels, wall > shelves, air vents? Have they also to be painted in dark colors or that > would be an overkill? I would appreciate if you share your solutions. > > Alexander > > > Alexander Jurkevic, PhD > Associate Director > Molecular Cytology Core > University of Missouri > 120 Life Sciences Center > 1201 E. Rollins St. > Columbia, MO 65211-7310 > > Phone: 573-882-4895 > Fax: 573-884-9676 > > > |
![]() ![]() |
Mark Cannell-2 |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** There is no need to go so far as to paint walls black if you eliminate light sources. A solid Faraday cage around the scope with a black curtain and interior is just as good and fixes the monitor light pollution problem. The room lighting should be easily dimmable from the scope/workstation. HTH Mark On 10/12/2013, at 12:40 am, Jurkevic, Aleksandr <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All, > > We are about to renovate the room for two-photon microscopy and I wonder what suggestions you might have about reducing light pollution that affects NDDs. As one of the measures, we are planning to paint walls and ceiling in black. What about "small" things like aluminum electrical channels, wall shelves, air vents? Have they also to be painted in dark colors or that would be an overkill? I would appreciate if you share your solutions. > > Alexander > > > Alexander Jurkevic, PhD > Associate Director > Molecular Cytology Core > University of Missouri > 120 Life Sciences Center > 1201 E. Rollins St. > Columbia, MO 65211-7310 > > Phone: 573-882-4895 > Fax: 573-884-9676 > > Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
![]() ![]() |
Pascal Weber |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Alexander, I install many two microscopes. Infact i works on plants and mice.I built a blac box around my upright microscope.It's very important to eliminate any light pollution. So i have many pictures of my rooms. My mail [hidden email] |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by Mark Cannell-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I agree its best if you can just box in the microscope itself: Then you don't have to worry about the room so much. In our case, however, we work with a lot of live samples that require a lot of checking on, and plumbing and heaters that don't work well with an enclosure, which is why we darkened the entire room. That way you have relatively free access to poke and prod your sample. If you don't need easy access, or don't mind constantly opening the enclosure to check on your sample, the box method works very well and is certainly simpler. If you are building a room from scratch though and have the option, I still recommend painting the walls a dark colour as the cost difference will be small compared to the traditional white or beige paint and will always give an advantage. Craig On Mon, Dec 9, 2013 at 6:44 PM, Mark Cannell <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > There is no need to go so far as to paint walls black if you eliminate > light sources. A solid Faraday cage around the scope with a black curtain > and interior is just as good and fixes the monitor light pollution > problem. The room lighting should be easily dimmable from the > scope/workstation. > > HTH > > Mark > > On 10/12/2013, at 12:40 am, Jurkevic, Aleksandr <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hello All, > > > > We are about to renovate the room for two-photon microscopy and I wonder > what suggestions you might have about reducing light pollution that affects > NDDs. As one of the measures, we are planning to paint walls and ceiling in > black. What about "small" things like aluminum electrical channels, wall > shelves, air vents? Have they also to be painted in dark colors or that > would be an overkill? I would appreciate if you share your solutions. > > > > Alexander > > > > > > Alexander Jurkevic, PhD > > Associate Director > > Molecular Cytology Core > > University of Missouri > > 120 Life Sciences Center > > 1201 E. Rollins St. > > Columbia, MO 65211-7310 > > > > Phone: 573-882-4895 > > Fax: 573-884-9676 > > > > > > Mark B. Cannell Ph.D. FRSNZ > Professor of Cardiac Cell Biology > School of Physiology & Pharmacology > Medical Sciences Building > University of Bristol > Bristol > BS8 1TD UK > > [hidden email] > |
![]() ![]() |
Mark Cannell-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ok, but thinking outside the box, the key is to eliminate/control sources. As long as the NDD can’t see the sources you don’t need a black room. I would prefer a big box around the scope -it could be big enough to let you fiddle but still block the monitor which is a major source of light ion the room. Even a single shield between the monitor and scope will work and then use mat dark green/blue walls and use a red filter over the monitor giving a much more pleasant work environment. HTH Cheers Mark On 10/12/2013, at 2:01 am, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I agree its best if you can just box in the microscope itself: Then you > don't have to worry about the room so much. In our case, however, we work > with a lot of live samples that require a lot of checking on, and plumbing > and heaters that don't work well with an enclosure, which is why we > darkened the entire room. That way you have relatively free access to poke > and prod your sample. If you don't need easy access, or don't mind > constantly opening the enclosure to check on your sample, the box method > works very well and is certainly simpler. If you are building a room from > scratch though and have the option, I still recommend painting the walls a > dark colour as the cost difference will be small compared to the > traditional white or beige paint and will always give an advantage. > > Craig > > > On Mon, Dec 9, 2013 at 6:44 PM, Mark Cannell <[hidden email]>wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> There is no need to go so far as to paint walls black if you eliminate >> light sources. A solid Faraday cage around the scope with a black curtain >> and interior is just as good and fixes the monitor light pollution >> problem. The room lighting should be easily dimmable from the >> scope/workstation. >> >> HTH >> >> Mark >> >> On 10/12/2013, at 12:40 am, Jurkevic, Aleksandr <[hidden email]> >> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hello All, >>> >>> We are about to renovate the room for two-photon microscopy and I wonder >> what suggestions you might have about reducing light pollution that affects >> NDDs. As one of the measures, we are planning to paint walls and ceiling in >> black. What about "small" things like aluminum electrical channels, wall >> shelves, air vents? Have they also to be painted in dark colors or that >> would be an overkill? I would appreciate if you share your solutions. >>> >>> Alexander >>> >>> >>> Alexander Jurkevic, PhD >>> Associate Director >>> Molecular Cytology Core >>> University of Missouri >>> 120 Life Sciences Center >>> 1201 E. Rollins St. >>> Columbia, MO 65211-7310 >>> >>> Phone: 573-882-4895 >>> Fax: 573-884-9676 >>> >>> >> >> Mark B. Cannell Ph.D. FRSNZ >> Professor of Cardiac Cell Biology >> School of Physiology & Pharmacology >> Medical Sciences Building >> University of Bristol >> Bristol >> BS8 1TD UK >> >> [hidden email] >> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Medical Sciences Building University of Bristol Bristol BS8 1TD UK [hidden email] |
![]() ![]() |
Pascal Weber |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I work primarily on the mouse. I installed my microscope with a dark part and a bright part (natural) to my surgeries. I had to create a black heated around my upright microscope box. I have all the references to build oneself. I also modify the plate to accept the sample. The walls were painted dark blue mat. But anyway your laser should never leave the safety casing. Know that dark walls absolutely no protection from parasites and dangerous reflections pulsed infrared lasers. My Systems: Zeiss Maitai. I'm building a 2P + OPO. Coherent. |
![]() ![]() |
Wilson, Sean (LLU) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We used the K.I.S.S. principle on our ziess 710 NLO. Based on empirical evidence, we ended up draping blackout cloth across the top of the scope to keep out ambient light from the monitor, ect. & turned off external light sources (i.e. EXFO). _____________________________________________ Sean M. Wilson, Ph.D. Assistant Professor, Center for Perinatal Biology Technical Director, Advanced Imaging and Microscopy Core Facility Loma Linda University School of Medicine Loma Linda, CA 92350 [hidden email] 909-558-4325 (Main Office) x45209 (Note new campus extension) x42183 (RW 1561 laboratory) 909-558-4029 (Fax) 909-647-7021 (Cell) http://microscopy.llu.edu Check out our video on the homepage! http://www.llu.edu/medicine/basic-sciences/faculty/pharmacology/wilson-sean.page? CONFIDENTIALITY NOTICE: This e-mail and any attachments are confidential and may also be privileged. If you are not the named recipient, please notify the sender immediately and delete the contents of this message without disclosing the contents to anyone, using them for any purpose, or storing or copying the information on any medium. ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Pascal Weber [[hidden email]] Sent: Monday, December 09, 2013 6:27 PM To: [hidden email] Subject: Re: room for two-photon microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I work primarily on the mouse. I installed my microscope with a dark part and a bright part (natural) to my surgeries. I had to create a black heated around my upright microscope box. I have all the references to build oneself. I also modify the plate to accept the sample. The walls were painted dark blue mat. But anyway your laser should never leave the safety casing. Know that dark walls absolutely no protection from parasites and dangerous reflections pulsed infrared lasers. My Systems: Zeiss Maitai. I'm building a 2P + OPO. Coherent. |
![]() ![]() |
Sylvie Le Guyader |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi there We have built low table-type things to raise the equipment that is normally on the floor like the 2P laser cooler and power supply. This way the fans do not suck the dust in and it is easier to keep the room clean. We have also built large cheap rectangular boxes that are placed just above the equipment that produces heat (2P cooler and power supply, argon laser power supply). These boxes are about 30 cm above the equipment and are connected to the normal ventilation system of the building. They channel the heat out very efficiently as there is a weakish but constant air suction. You can avoid costly ventilation work that way. On top of that we have an air con unit set to 23deg Celsius so it is quite comfortable to work in the room. The best way is of course to have a heat and noise producing equipment in an adjacent room and a hole in the wall for the cables. We have a full size incubator on which we can add black panels when we use the NDDs or the GaAsP. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet Hälsovägen 7 Novum, G lift, floor 6 14157 Huddinge Sweden office: +46 (0) 8 5248 1107 LCI room 1: +46 (0) 8 5248 1172 LCI room 2: +46 (0) 8 5248 3542 mobile: +46 (0) 73 733 5008 > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Jurkevic, > Aleksandr > Sent: 10 December 2013 01:40 > To: [hidden email] > Subject: room for two-photon microscopy > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All, > > We are about to renovate the room for two-photon microscopy and I wonder what > suggestions you might have about reducing light pollution that affects NDDs. As > one of the measures, we are planning to paint walls and ceiling in black. What > about "small" things like aluminum electrical channels, wall shelves, air vents? Have > they also to be painted in dark colors or that would be an overkill? I would > appreciate if you share your solutions. > > Alexander > > > Alexander Jurkevic, PhD > Associate Director > Molecular Cytology Core > University of Missouri > 120 Life Sciences Center > 1201 E. Rollins St. > Columbia, MO 65211-7310 > > Phone: 573-882-4895 > Fax: 573-884-9676 > > |
![]() ![]() |
Watkins, Simon C |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Apart from the endless idiot lights that need to be taped over the monitor is the biggest issue. The best solution in our hands has been to use security film (3M) which makes the monitor a little dimmer straight on but effectively removes all light at an angle and removes the need for draping curtains etc. On our second system we have built the proverbial black plexiglass box, which provides heat and allows room lightŠ it is certainly the best solution all round, I can send pics and design details if anyone wants them (actually you can see it if you look under Nikon multi photon in the AAA section of our resource page at www.cbi.pitt.edu). S. Simon Watkins Ph.D Professor and Vice Chair Cell Biology Professor Immunology Director Center for Biologic Imaging University of Pittsburgh Bsts 225 3550 terrace st Pittsburgh PA 15261 Www.cbi.pitt.edu <http://Www.cbi.pitt.edu/> 412-352-2277 On 12/9/13, 7:40 PM, "Jurkevic, Aleksandr" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello All, > >We are about to renovate the room for two-photon microscopy and I wonder >what suggestions you might have about reducing light pollution that >affects NDDs. As one of the measures, we are planning to paint walls and >ceiling in black. What about "small" things like aluminum electrical >channels, wall shelves, air vents? Have they also to be painted in dark >colors or that would be an overkill? I would appreciate if you share your >solutions. > >Alexander > > >Alexander Jurkevic, PhD >Associate Director >Molecular Cytology Core >University of Missouri >120 Life Sciences Center >1201 E. Rollins St. >Columbia, MO 65211-7310 > >Phone: 573-882-4895 >Fax: 573-884-9676 > > |
![]() ![]() |
Tobias Rose |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We usually use a dark cage as well. However, since we use monitors for visual stimulation of living animals we also synchronize the LED backlight of the screens to the turnaround phase of our resonant scanners. This way, the screens (both for stimulation and in some case also all remaining screens in the room) are only bright during the ~30% of scanning time where we reject the acquired data anyways. Works like a charm. Tobias (Leinweber et al. 2013, JOVE, in press) -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jurkevic, Aleksandr Sent: Tuesday, December 10, 2013 01:40 To: [hidden email] Subject: room for two-photon microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello All, We are about to renovate the room for two-photon microscopy and I wonder what suggestions you might have about reducing light pollution that affects NDDs. As one of the measures, we are planning to paint walls and ceiling in black. What about "small" things like aluminum electrical channels, wall shelves, air vents? Have they also to be painted in dark colors or that would be an overkill? I would appreciate if you share your solutions. Alexander Alexander Jurkevic, PhD Associate Director Molecular Cytology Core University of Missouri 120 Life Sciences Center 1201 E. Rollins St. Columbia, MO 65211-7310 Phone: 573-882-4895 Fax: 573-884-9676 |
![]() ![]() |
Arvydas Matiukas |
![]() |
In reply to this post by Jurkevic, Aleksandr
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello List, We are going to move to new research building with all current microscopy equipment. Few new microscopes ( confocal, 2-photon and STED-superresolution) will be installed as well. I ordered all the walls painted matte black. I hope it makes a lot of difference . You already can notice that when making a photo. We installed red overhead light as an selectable option. Taking the occasion of the thread I would like to ask about other important issues to follow up before we move. Manufacturers of microscopes suggest to check the room for vibration, conditioning airflow, dust-free, clean power. Any feedback/advise is highly appreciated. Best regards, Arvydas Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core Department of Neurosci& Physiology SUNY Upstate Medical University 766 Irving Ave., WH 3167 Syracuse, NY 13210 tel.: 315-464-7997 fax: 315-464-8014 email: [hidden email] >>> "Jurkevic, Aleksandr" <[hidden email]> 12/9/2013 7:40 PM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello All, We are about to renovate the room for two-photon microscopy and I wonder what suggestions you might have about reducing light pollution that affects NDDs. As one of the measures, we are planning to paint walls and ceiling in black. What about "small" things like aluminum electrical channels, wall shelves, air vents? Have they also to be painted in dark colors or that would be an overkill? I would appreciate if you share your solutions. Alexander Alexander Jurkevic, PhD Associate Director Molecular Cytology Core University of Missouri 120 Life Sciences Center 1201 E. Rollins St. Columbia, MO 65211-7310 Phone: 573-882-4895 Fax: 573-884-9676 |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You will need cooling to handle the heat from the lamps and lasers, as well as the people occupying the room. You will want filtered air to keep dust to a minimum. Temperature should be stable to +/-1 degree C, which is often one of the hardest criteria to meet. Humidity should be kept reasonable, and provision should be made for dehumidification if it is a damp environment. Regarding vibration, is the room on an upper floor? Basement is ideal, but if you are on an upper floor you will want good passively-damped tables that are fairly thick. Also make sure you can get the optical tables into place without having to knock a hole in the wall, so check door widths, elevator capacities, etc. Make sure light switches for the room are near where the operator(s) will sit. Make sure sample prep areas are nearby and appropriately equipped (sinks, incubators, whatever you need). Make sure you have mounts for any gas bottles and that they meet the safety standards required in your area. That's what I've got off the top of my head, I'm sure others will have much more to add! Craig On Wed, Dec 11, 2013 at 3:02 PM, Arvydas Matiukas <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello List, > > We are going to move to new research building with all > current microscopy equipment. Few new microscopes ( > confocal, 2-photon and STED-superresolution) > will be installed as well. I ordered all the walls painted > matte black. I hope it makes a lot of difference . You > already can notice that when making a photo. > We installed red overhead light as an selectable option. > > Taking the occasion of the thread I would like to ask > about other important issues to follow up before we > move. Manufacturers of microscopes suggest to check > the room for vibration, conditioning airflow, dust-free, > clean power. Any feedback/advise is highly appreciated. > > Best regards, > Arvydas > > > > Arvydas Matiukas, Ph.D. > Director of Confocal&Two-Photon Core > Department of Neurosci& Physiology > SUNY Upstate Medical University > 766 Irving Ave., WH 3167 > Syracuse, NY 13210 > tel.: 315-464-7997 > fax: 315-464-8014 > email: [hidden email] > >>> "Jurkevic, Aleksandr" <[hidden email]> 12/9/2013 7:40 PM >>> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All, > > We are about to renovate the room for two-photon microscopy and I wonder > what suggestions you might have about reducing light pollution that affects > NDDs. As one of the measures, we are planning to paint walls and ceiling in > black. What about "small" things like aluminum electrical channels, wall > shelves, air vents? Have they also to be painted in dark colors or that > would be an overkill? I would appreciate if you share your solutions. > > Alexander > > > Alexander Jurkevic, PhD > Associate Director > Molecular Cytology Core > University of Missouri > 120 Life Sciences Center > 1201 E. Rollins St. > Columbia, MO 65211-7310 > > Phone: 573-882-4895 > Fax: 573-884-9676 > > > |
![]() |
In reply to this post by Arvydas Matiukas
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** If you are using Isoflurane or other gas based anesthesia, make sure the room is well ventilated even when you have a charcoal filter in the microscope's environmental chamber. =========================================================================== Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU Langone Medical Center Cell: 914-309-3270 Lab: 212-263-3208 http://ocs.med.nyu.edu/microscopy & http://www.med.nyu.edu/skirball-lab/dustinlab/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Arvydas Matiukas Sent: Wednesday, December 11, 2013 5:03 PM To: [hidden email] Subject: Re: room for two-photon microscopy: checking environment ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello List, We are going to move to new research building with all current microscopy equipment. Few new microscopes ( confocal, 2-photon and STED-superresolution) will be installed as well. I ordered all the walls painted matte black. I hope it makes a lot of difference . You already can notice that when making a photo. We installed red overhead light as an selectable option. Taking the occasion of the thread I would like to ask about other important issues to follow up before we move. Manufacturers of microscopes suggest to check the room for vibration, conditioning airflow, dust-free, clean power. Any feedback/advise is highly appreciated. Best regards, Arvydas Arvydas Matiukas, Ph.D. Director of Confocal&Two-Photon Core Department of Neurosci& Physiology SUNY Upstate Medical University 766 Irving Ave., WH 3167 Syracuse, NY 13210 tel.: 315-464-7997 fax: 315-464-8014 email: [hidden email] >>> "Jurkevic, Aleksandr" <[hidden email]> 12/9/2013 7:40 PM >>> ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello All, We are about to renovate the room for two-photon microscopy and I wonder what suggestions you might have about reducing light pollution that affects NDDs. As one of the measures, we are planning to paint walls and ceiling in black. What about "small" things like aluminum electrical channels, wall shelves, air vents? Have they also to be painted in dark colors or that would be an overkill? I would appreciate if you share your solutions. Alexander Alexander Jurkevic, PhD Associate Director Molecular Cytology Core University of Missouri 120 Life Sciences Center 1201 E. Rollins St. Columbia, MO 65211-7310 Phone: 573-882-4895 Fax: 573-884-9676 ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** To add to the ventilation comment, we had a flexible tube installed near the microscope connected to a filtered exhaust. When we are using isoflurane we can place the open end of the tube near the stage. It has a mild draw on it, not strong enough to disrupt the environment, but strong enough to pull the gas away from the user. Craig Brideau On Thu, Dec 19, 2013 at 10:48 AM, Cammer, Michael < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > If you are using Isoflurane or other gas based anesthesia, make sure the > room is well ventilated even when you have a charcoal filter in the > microscope's environmental chamber. > > =========================================================================== > Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU > Langone Medical Center > Cell: 914-309-3270 Lab: 212-263-3208 > http://ocs.med.nyu.edu/microscopy & > http://www.med.nyu.edu/skirball-lab/dustinlab/ > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Arvydas Matiukas > Sent: Wednesday, December 11, 2013 5:03 PM > To: [hidden email] > Subject: Re: room for two-photon microscopy: checking environment > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello List, > > We are going to move to new research building with all current microscopy > equipment. Few new microscopes ( confocal, 2-photon and > STED-superresolution) will be installed as well. I ordered all the walls > painted matte black. I hope it makes a lot of difference . You already can > notice that when making a photo. > We installed red overhead light as an selectable option. > > Taking the occasion of the thread I would like to ask about other > important issues to follow up before we move. Manufacturers of microscopes > suggest to check the room for vibration, conditioning airflow, dust-free, > clean power. Any feedback/advise is highly appreciated. > > Best regards, > Arvydas > > > > Arvydas Matiukas, Ph.D. > Director of Confocal&Two-Photon Core > Department of Neurosci& Physiology > SUNY Upstate Medical University > 766 Irving Ave., WH 3167 > Syracuse, NY 13210 > tel.: 315-464-7997 > fax: 315-464-8014 > email: [hidden email] > >>> "Jurkevic, Aleksandr" <[hidden email]> 12/9/2013 7:40 PM >>> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello All, > > We are about to renovate the room for two-photon microscopy and I wonder > what suggestions you might have about reducing light pollution that affects > NDDs. As one of the measures, we are planning to paint walls and ceiling in > black. What about "small" things like aluminum electrical channels, wall > shelves, air vents? Have they also to be painted in dark colors or that > would be an overkill? I would appreciate if you share your solutions. > > Alexander > > > Alexander Jurkevic, PhD > Associate Director > Molecular Cytology Core > University of Missouri > 120 Life Sciences Center > 1201 E. Rollins St. > Columbia, MO 65211-7310 > > Phone: 573-882-4895 > Fax: 573-884-9676 > > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the > intended recipient(s) and may contain information that is proprietary, > confidential, and exempt from disclosure under applicable law. Any > unauthorized review, use, disclosure, or distribution is prohibited. If you > have received this email in error please notify the sender by return email > and delete the original message. Please note, the recipient should check > this email and any attachments for the presence of viruses. The > organization accepts no liability for any damage caused by any virus > transmitted by this email. > ================================= > |