spectral versus well filtered in plants

classic Classic list List threaded Threaded
♦
♦
8 messages Options
Christian-103 Christian-103
Reply | Threaded
Open this post in threaded view
♦
♦
|

spectral versus well filtered in plants

Recently a new faculty member has introduced new constructs of both YFP and CFP for localization in plant cells, mostly tobacco and arabidopsis.  Our current system, a FluoView 500 is not set up for this work, so I've been asked to investigate new avenues.

In any case, I think we've all seen a spectral system separate GFP, Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP colocalization in plant cells would a spectral system offer more utility than a system with better filter sets and laser lines?

Obviously there is a cost question in relation to utility here.  Also, here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon systems.  If anyone would like comment on those in particular, that'd be great.

I also have a more specific question, with several groups targeting chloroplasts, I should probably just lean towards a spectral system?  Finally, do YFP and GFP overlap too greatly in plants (pH difference?!?) to be separated by either system?

If any has some negative feedback, or suggestions, please feel free to email me privately.  I find folks are awfully nice on the list, but when we're talking several hundred thousand dollars, I need brutal honesty.

Thank you.

Christian

Kurt Thorn Kurt Thorn
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

You won't be able to separate GFP and YFP reliably without a spectral
system; the emission wavelengths aren't separated by enough to do this
with conventional filters.

There are other fluorescent protein combinations you could consider
which might be better behaved, such as one of the new BFPs or a
UV-excited GFP like Sapphire which would let you do 4 proteins with
better separation between channels.  I think you can also multiplex CFP,
GFP, mkOrange, and a far red FP like TagFP635 or tHcRed.  You could
probably add BFP or Sapphire to that mix to do 5 proteins, although you
will probably start getting some crosstalk you need to deal with.

Kurt

Christian wrote:

>
> Recently a new faculty member has introduced new constructs of both
> YFP and CFP for localization in plant cells, mostly tobacco and
> arabidopsis.  Our current system, a FluoView 500 is not set up for
> this work, so I've been asked to investigate new avenues.
>
> In any case, I think we've all seen a spectral system separate GFP,
> Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
> colocalization in plant cells would a spectral system offer more
> utility than a system with better filter sets and laser lines?
>
> Obviously there is a cost question in relation to utility here.  Also,
> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
> systems.  If anyone would like comment on those in particular, that'd
> be great.
>
> I also have a more specific question, with several groups targeting
> chloroplasts, I should probably just lean towards a spectral system?  
> Finally, do YFP and GFP overlap too greatly in plants (pH
> difference?!?) to be separated by either system?
>
> If any has some negative feedback, or suggestions, please feel free to
> email me privately.  I find folks are awfully nice on the list, but
> when we're talking several hundred thousand dollars, I need brutal
> honesty.
>
> Thank you.
>
> Christian
>


--
Kurt Thorn, PhD
Director, Nikon Imaging Center
University of California San Francisco

UCSF MC 2140
Genentech Hall Room S252
600 16th St.
San Francisco, CA 94158-2517

http://nic.ucsf.edu
phone 415.514.9709
fax   415.514.4300
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

As a general note, my limited experience with spectral imaging is that
unless you have a strong signal, the image quality of the unmixed spectra
can be problematic at best.  Of course, it could be the system operator.  If
you demo any system, make sure you use one of your own preps, not a
commercial slide.

C

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "Kurt Thorn" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, April 02, 2009 10:13 AM
Subject: Re: spectral versus well filtered in plants


> You won't be able to separate GFP and YFP reliably without a spectral
> system; the emission wavelengths aren't separated by enough to do this
> with conventional filters.
>
> There are other fluorescent protein combinations you could consider which
> might be better behaved, such as one of the new BFPs or a UV-excited GFP
> like Sapphire which would let you do 4 proteins with better separation
> between channels.  I think you can also multiplex CFP, GFP, mkOrange, and
> a far red FP like TagFP635 or tHcRed.  You could probably add BFP or
> Sapphire to that mix to do 5 proteins, although you will probably start
> getting some crosstalk you need to deal with.
>
> Kurt
>
> Christian wrote:
>>
>> Recently a new faculty member has introduced new constructs of both YFP
>> and CFP for localization in plant cells, mostly tobacco and arabidopsis.
>> Our current system, a FluoView 500 is not set up for this work, so I've
>> been asked to investigate new avenues.
>>
>> In any case, I think we've all seen a spectral system separate GFP, Sytox
>> Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
>> colocalization in plant cells would a spectral system offer more utility
>> than a system with better filter sets and laser lines?
>>
>> Obviously there is a cost question in relation to utility here.  Also,
>> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
>> systems.  If anyone would like comment on those in particular, that'd be
>> great.
>>
>> I also have a more specific question, with several groups targeting
>> chloroplasts, I should probably just lean towards a spectral system?
>> Finally, do YFP and GFP overlap too greatly in plants (pH difference?!?)
>> to be separated by either system?
>>
>> If any has some negative feedback, or suggestions, please feel free to
>> email me privately.  I find folks are awfully nice on the list, but when
>> we're talking several hundred thousand dollars, I need brutal honesty.
>>
>> Thank you.
>>
>> Christian
>>
>
>
> --
> Kurt Thorn, PhD
> Director, Nikon Imaging Center
> University of California San Francisco
>
> UCSF MC 2140
> Genentech Hall Room S252
> 600 16th St.
> San Francisco, CA 94158-2517
>
> http://nic.ucsf.edu
> phone 415.514.9709
> fax   415.514.4300
>
Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

Hello, I think that you would not be happy if you try to separate these
spectra by using filters. You will need a spectral detector. Both Zeiss
and Leica make very good spectral Confocal microscopes that will do the
job nicely. You may also want to consider the Nuance system from CRI for
around $80K US (no commercial interest). Cheers,
http://www.cri-inc.com/products/nuance.asp


Brian Armstrong PhD
Manager, Light Microscopy Core
Beckman Research Institute
1450 East Duarte Rd
Duarte, CA 91010
626-256-4673 x62872
http://www.cityofhope.org/SharedResources/LightMicroscopy/LightMicroHome
.htm
 
 
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Carl Boswell
Sent: Thursday, April 02, 2009 2:07 PM
To: [hidden email]
Subject: Re: spectral versus well filtered in plants

As a general note, my limited experience with spectral imaging is that
unless you have a strong signal, the image quality of the unmixed
spectra
can be problematic at best.  Of course, it could be the system operator.
If
you demo any system, make sure you use one of your own preps, not a
commercial slide.

C

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "Kurt Thorn" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, April 02, 2009 10:13 AM
Subject: Re: spectral versus well filtered in plants


> You won't be able to separate GFP and YFP reliably without a spectral
> system; the emission wavelengths aren't separated by enough to do this

> with conventional filters.
>
> There are other fluorescent protein combinations you could consider
which
> might be better behaved, such as one of the new BFPs or a UV-excited
GFP
> like Sapphire which would let you do 4 proteins with better separation

> between channels.  I think you can also multiplex CFP, GFP, mkOrange,
and
> a far red FP like TagFP635 or tHcRed.  You could probably add BFP or
> Sapphire to that mix to do 5 proteins, although you will probably
start
> getting some crosstalk you need to deal with.
>
> Kurt
>
> Christian wrote:
>>
>> Recently a new faculty member has introduced new constructs of both
YFP
>> and CFP for localization in plant cells, mostly tobacco and
arabidopsis.
>> Our current system, a FluoView 500 is not set up for this work, so
I've
>> been asked to investigate new avenues.
>>
>> In any case, I think we've all seen a spectral system separate GFP,
Sytox
>> Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
>> colocalization in plant cells would a spectral system offer more
utility
>> than a system with better filter sets and laser lines?
>>
>> Obviously there is a cost question in relation to utility here.
Also,
>> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon

>> systems.  If anyone would like comment on those in particular, that'd
be
>> great.
>>
>> I also have a more specific question, with several groups targeting
>> chloroplasts, I should probably just lean towards a spectral system?
>> Finally, do YFP and GFP overlap too greatly in plants (pH
difference?!?)
>> to be separated by either system?
>>
>> If any has some negative feedback, or suggestions, please feel free
to
>> email me privately.  I find folks are awfully nice on the list, but
when
>> we're talking several hundred thousand dollars, I need brutal
honesty.

>>
>> Thank you.
>>
>> Christian
>>
>
>
> --
> Kurt Thorn, PhD
> Director, Nikon Imaging Center
> University of California San Francisco
>
> UCSF MC 2140
> Genentech Hall Room S252
> 600 16th St.
> San Francisco, CA 94158-2517
>
> http://nic.ucsf.edu
> phone 415.514.9709
> fax   415.514.4300
>


---------------------------------------------------------------------
 
SECURITY/CONFIDENTIALITY WARNING:  
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.  
---------------------------------------------------------------------
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

In reply to this post by Boswell, Carl A - (cboswell)
We use a Nikon spectral system.  Their detector array is photomultiplier-based so it is quite sensitive for a spectral system.  We use YFP and GFP and it gives very good results.

Craig Brideau
Hotchkiss Brain Institute
University of Calgary, Clinical Neuroscience
Calgary, AB
Canada

On Thu, Apr 2, 2009 at 3:06 PM, Carl Boswell <[hidden email]> wrote:
As a general note, my limited experience with spectral imaging is that unless you have a strong signal, the image quality of the unmixed spectra can be problematic at best.  Of course, it could be the system operator.  If you demo any system, make sure you use one of your own preps, not a commercial slide.

C

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message ----- From: "Kurt Thorn" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, April 02, 2009 10:13 AM
Subject: Re: spectral versus well filtered in plants



You won't be able to separate GFP and YFP reliably without a spectral system; the emission wavelengths aren't separated by enough to do this with conventional filters.

There are other fluorescent protein combinations you could consider which might be better behaved, such as one of the new BFPs or a UV-excited GFP like Sapphire which would let you do 4 proteins with better separation between channels.  I think you can also multiplex CFP, GFP, mkOrange, and a far red FP like TagFP635 or tHcRed.  You could probably add BFP or Sapphire to that mix to do 5 proteins, although you will probably start getting some crosstalk you need to deal with.

Kurt

Christian wrote:

Recently a new faculty member has introduced new constructs of both YFP and CFP for localization in plant cells, mostly tobacco and arabidopsis. Our current system, a FluoView 500 is not set up for this work, so I've been asked to investigate new avenues.

In any case, I think we've all seen a spectral system separate GFP, Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP colocalization in plant cells would a spectral system offer more utility than a system with better filter sets and laser lines?

Obviously there is a cost question in relation to utility here.  Also, here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon systems.  If anyone would like comment on those in particular, that'd be great.

I also have a more specific question, with several groups targeting chloroplasts, I should probably just lean towards a spectral system? Finally, do YFP and GFP overlap too greatly in plants (pH difference?!?) to be separated by either system?

If any has some negative feedback, or suggestions, please feel free to email me privately.  I find folks are awfully nice on the list, but when we're talking several hundred thousand dollars, I need brutal honesty.

Thank you.

Christian



--
Kurt Thorn, PhD
Director, Nikon Imaging Center
University of California San Francisco

UCSF MC 2140
Genentech Hall Room S252
600 16th St.
San Francisco, CA 94158-2517

http://nic.ucsf.edu
phone 415.514.9709
fax   415.514.4300


Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

In reply to this post by Kurt Thorn
All true.  We managed to separate CFP, YFP, and GFP plus chlorophyll and
cell walls (blue autofluorescence from 405 excitation) in one sample on a
spectral system (SP2).  It was a bit of an effort, but is possible.  If you
add bi-directional scanning it becomes easier, too.  I can't imagine doing
this without being able to adjust the emission wavebands collected.
cheers,
Rosemary

Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph 61 2 6246 5475
fx 61 2 6246 5334


On 3/04/09 4:13 AM, "Kurt Thorn" <[hidden email]> wrote:

> You won't be able to separate GFP and YFP reliably without a spectral
> system; the emission wavelengths aren't separated by enough to do this
> with conventional filters.
>
> There are other fluorescent protein combinations you could consider
> which might be better behaved, such as one of the new BFPs or a
> UV-excited GFP like Sapphire which would let you do 4 proteins with
> better separation between channels.  I think you can also multiplex CFP,
> GFP, mkOrange, and a far red FP like TagFP635 or tHcRed.  You could
> probably add BFP or Sapphire to that mix to do 5 proteins, although you
> will probably start getting some crosstalk you need to deal with.
>
> Kurt
>
> Christian wrote:
>>
>> Recently a new faculty member has introduced new constructs of both
>> YFP and CFP for localization in plant cells, mostly tobacco and
>> arabidopsis.  Our current system, a FluoView 500 is not set up for
>> this work, so I've been asked to investigate new avenues.
>>
>> In any case, I think we've all seen a spectral system separate GFP,
>> Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
>> colocalization in plant cells would a spectral system offer more
>> utility than a system with better filter sets and laser lines?
>>
>> Obviously there is a cost question in relation to utility here.  Also,
>> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
>> systems.  If anyone would like comment on those in particular, that'd
>> be great.
>>
>> I also have a more specific question, with several groups targeting
>> chloroplasts, I should probably just lean towards a spectral system?
>> Finally, do YFP and GFP overlap too greatly in plants (pH
>> difference?!?) to be separated by either system?
>>
>> If any has some negative feedback, or suggestions, please feel free to
>> email me privately.  I find folks are awfully nice on the list, but
>> when we're talking several hundred thousand dollars, I need brutal
>> honesty.
>>
>> Thank you.
>>
>> Christian
>>
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

Spectral unmixing software is also invaluable.  We have software in our lab that can handle data from the Nikon system and perform spectral unmixing on it.  It's open source too: ImageTrak

Craig


On Thu, Apr 2, 2009 at 4:40 PM, Rosemary White <[hidden email]> wrote:
All true.  We managed to separate CFP, YFP, and GFP plus chlorophyll and
cell walls (blue autofluorescence from 405 excitation) in one sample on a
spectral system (SP2).  It was a bit of an effort, but is possible.  If you
add bi-directional scanning it becomes easier, too.  I can't imagine doing
this without being able to adjust the emission wavebands collected.
cheers,
Rosemary

Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

ph 61 2 6246 5475
fx 61 2 6246 5334


On 3/04/09 4:13 AM, "Kurt Thorn" <[hidden email]> wrote:

> You won't be able to separate GFP and YFP reliably without a spectral
> system; the emission wavelengths aren't separated by enough to do this
> with conventional filters.
>
> There are other fluorescent protein combinations you could consider
> which might be better behaved, such as one of the new BFPs or a
> UV-excited GFP like Sapphire which would let you do 4 proteins with
> better separation between channels.  I think you can also multiplex CFP,
> GFP, mkOrange, and a far red FP like TagFP635 or tHcRed.  You could
> probably add BFP or Sapphire to that mix to do 5 proteins, although you
> will probably start getting some crosstalk you need to deal with.
>
> Kurt
>
> Christian wrote:
>>
>> Recently a new faculty member has introduced new constructs of both
>> YFP and CFP for localization in plant cells, mostly tobacco and
>> arabidopsis.  Our current system, a FluoView 500 is not set up for
>> this work, so I've been asked to investigate new avenues.
>>
>> In any case, I think we've all seen a spectral system separate GFP,
>> Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
>> colocalization in plant cells would a spectral system offer more
>> utility than a system with better filter sets and laser lines?
>>
>> Obviously there is a cost question in relation to utility here.  Also,
>> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
>> systems.  If anyone would like comment on those in particular, that'd
>> be great.
>>
>> I also have a more specific question, with several groups targeting
>> chloroplasts, I should probably just lean towards a spectral system?
>> Finally, do YFP and GFP overlap too greatly in plants (pH
>> difference?!?) to be separated by either system?
>>
>> If any has some negative feedback, or suggestions, please feel free to
>> email me privately.  I find folks are awfully nice on the list, but
>> when we're talking several hundred thousand dollars, I need brutal
>> honesty.
>>
>> Thank you.
>>
>> Christian
>>
>

O'Toole, P O'Toole, P
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: spectral versus well filtered in plants

In reply to this post by Christian-103
Dear Christian

To add to other comments. When using CFP and YFP in plant materials, it
can sometimes be better to use the 458nm laser line in preference to the
405 nm line. Despite the increased excitation of YFP from the 458, the
458 nm line helps to minimise autofluorescence in the plant material
that can mask low levels of CFP expression.  Still not perfect, but this
has helped when using Chameleon and other CFP experiments.

When spectral profiling CFP and YFP with autofluorescence, you may not
need to excite the YFP with the 514 nm line if the expression levels are
similar or greater than that of CFP. This will help balance the
intensities so that one probe does not dominate the spectral profiles.

Good luck

Peter

Christian wrote:

>
> Recently a new faculty member has introduced new constructs of both
> YFP and CFP for localization in plant cells, mostly tobacco and
> arabidopsis.  Our current system, a FluoView 500 is not set up for
> this work, so I've been asked to investigate new avenues.
>
> In any case, I think we've all seen a spectral system separate GFP,
> Sytox Green and FITC in the same cell, but for CFP, YFP, GFP, RFP
> colocalization in plant cells would a spectral system offer more
> utility than a system with better filter sets and laser lines?
>
> Obviously there is a cost question in relation to utility here.  Also,
> here in Midwest of the US, we have primarily Zeiss, Olympus and Nikon
> systems.  If anyone would like comment on those in particular, that'd
> be great.
>
> I also have a more specific question, with several groups targeting
> chloroplasts, I should probably just lean towards a spectral system?  
> Finally, do YFP and GFP overlap too greatly in plants (pH
> difference?!?) to be separated by either system?
>
> If any has some negative feedback, or suggestions, please feel free to
> email me privately.  I find folks are awfully nice on the list, but
> when we're talking several hundred thousand dollars, I need brutal
> honesty.
>
> Thank you.
>
> Christian
>

--