Dear all,

In our facility, we are looking to purchase a stage top incubator in
order to do short term live cell imaging (5-30 min). We found two models
that look compact (similar size of a wellplate) and user friendly: the
Okolab H301-MINI and the Tokai Hit WELSX.

Does any anyone has positive or negative experience with one of those?


Any feedback to help us make a decision will be highly appreciated.

Thanks in advance,

Luigi


--
Luigi Bozzo, PhD
Bioimaging and Optics Platform (PT-BIOP)
Ecole Polytechnique Fédérale de Lausanne (EPFL)
Faculty of Life Sciences
Station 15, AI 0140
CH-1015 Lausanne

Phone: +41 21 693 9629
http://biop.epfl.ch/
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

RES: stage top incubator

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Luigi,

We have a Tokai Hit stage top incubator that has performed very well for the last 5 years.

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Bozzo Luigi
Enviada em: terça-feira, 23 de agosto de 2016 08:57
Para: [hidden email]
Assunto: stage top incubator

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

In our facility, we are looking to purchase a stage top incubator in order to do short term live cell imaging (5-30 min). We found two models that look compact (similar size of a wellplate) and user friendly: the Okolab H301-MINI and the Tokai Hit WELSX.

Does any anyone has positive or negative experience with one of those?


Any feedback to help us make a decision will be highly appreciated.

Thanks in advance,

Luigi


--
Luigi Bozzo, PhD
Bioimaging and Optics Platform (PT-BIOP) Ecole Polytechnique Fédérale de Lausanne (EPFL) Faculty of Life Sciences Station 15, AI 0140
CH-1015 Lausanne

Phone: +41 21 693 9629
http://biop.epfl.ch/
Dan Focht Dan Focht
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: stage top incubator (commercial response)

In reply to this post by Bozzo Luigi
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Luigi

I would like to address you micro-environmental control needs from a different approach.
Bioptechs does not make “stage-top box” type incubators.  
We recognize that live cell imaging environmental control is not a trivial matter.
Therefore, we specialize in specimen control not box control.
Our systems are available for immediate delivery and are designed to meet the specific requirements of each customer.

We make sure our systems are appropriate for your exacting needs including:
High magnification, high N.A. compatibility
Speed and uniformity of temperature control (RT to 37 in under two minutes)
Accurate objective temperature control regulated at the specimen plane not a heating collar.
Ability to accommodate nearly any specimen type on any scope upright or inverted
A user definable optical cavity
Gas and humidity control
Perfusion if necessary
Positional repeatability
Stability in the Z axis plane

Additional information on our systems is available at www.bioptechs.com.

Dan

> On Aug 23, 2016, at 7:56 AM, Bozzo Luigi <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all,
>
> In our facility, we are looking to purchase a stage top incubator in order to do short term live cell imaging (5-30 min). We found two models that look compact (similar size of a wellplate) and user friendly: the Okolab H301-MINI and the Tokai Hit WELSX.
>
> Does any anyone has positive or negative experience with one of those?
>
>
> Any feedback to help us make a decision will be highly appreciated.
>
> Thanks in advance,
>
> Luigi
>
>
> --
> Luigi Bozzo, PhD
> Bioimaging and Optics Platform (PT-BIOP)
> Ecole Polytechnique Fédérale de Lausanne (EPFL)
> Faculty of Life Sciences
> Station 15, AI 0140
> CH-1015 Lausanne
>
> Phone: +41 21 693 9629
> http://biop.epfl.ch/

Dan Focht
Bioptechs, Inc.
3560 Beck Rd.
Butler PA 16002
V724-282-7145
F724-282-0745
Toll Free 877 lIVE-CELL (548-3235)
[hidden email]
mmodel mmodel
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: stage top incubator

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We use LiveCell made by Pathology Devices. It works well and they have very good support.

Mike

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara
Sent: Tuesday, August 23, 2016 8:40 AM
To: [hidden email]
Subject: RES: stage top incubator

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Luigi,

We have a Tokai Hit stage top incubator that has performed very well for the last 5 years.

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Bozzo Luigi Enviada em: terça-feira, 23 de agosto de 2016 08:57
Para: [hidden email]
Assunto: stage top incubator

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

In our facility, we are looking to purchase a stage top incubator in order to do short term live cell imaging (5-30 min). We found two models that look compact (similar size of a wellplate) and user friendly: the Okolab H301-MINI and the Tokai Hit WELSX.

Does any anyone has positive or negative experience with one of those?


Any feedback to help us make a decision will be highly appreciated.

Thanks in advance,

Luigi


--
Luigi Bozzo, PhD
Bioimaging and Optics Platform (PT-BIOP) Ecole Polytechnique Fédérale de Lausanne (EPFL) Faculty of Life Sciences Station 15, AI 0140
CH-1015 Lausanne

Phone: +41 21 693 9629
http://biop.epfl.ch/
Cesare -ALEMBIC- Cesare -ALEMBIC-
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: RES: stage top incubator

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear Luigi,
we have many Oko-Lab devices in our facilicity (since 2005), small and bigger stage-incubators, cage incubators and so on.
All the systems have always worked well and we always found a good support for any problem and also honest and useful help for the choice of the most suitable device for our purpose/instruments/budget (very often with personal customisations, if necessary)

All the best,

Cesare


Dr. Cesare Covino
Facility Manager
ALEMBIC
Advanced Light and Electron Microscopy BioImaging Center
San Raffaele Scientific Institute
DIBIT 1  S2B4a
via Olgettina 58, 20132 - Milano - Italy
Tel +39-022643-4636(others -4640, -4641)
Fax +39-022643-4646
e-mail: [hidden email]<mailto:[hidden email]>
ALEMBIC webpage: www.hsr.it/research/alembic<http://www.hsr.it/research/alembic>


Il 23/08/2016 14:39, Renato Mortara ha scritto:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Luigi,

We have a Tokai Hit stage top incubator that has performed very well for the last 5 years.

Renato

-----Mensagem original-----
De: Confocal Microscopy List [mailto:[hidden email]] Em nome de Bozzo Luigi
Enviada em: terça-feira, 23 de agosto de 2016 08:57
Para: [hidden email]<mailto:[hidden email]>
Assunto: stage top incubator

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

In our facility, we are looking to purchase a stage top incubator in order to do short term live cell imaging (5-30 min). We found two models that look compact (similar size of a wellplate) and user friendly: the Okolab H301-MINI and the Tokai Hit WELSX.

Does any anyone has positive or negative experience with one of those?


Any feedback to help us make a decision will be highly appreciated.

Thanks in advance,

Luigi


--
Luigi Bozzo, PhD
Bioimaging and Optics Platform (PT-BIOP) Ecole Polytechnique Fédérale de Lausanne (EPFL) Faculty of Life Sciences Station 15, AI 0140
CH-1015 Lausanne

Phone: +41 21 693 9629
http://biop.epfl.ch/
.



<http://www.hsr.it/research/alembic>


Rispetta l’ambiente: non stampare questa mail se non è necessario.
Respect the environment: if it's not necessary, don't print this mail.

IL TUO 5XMILLE AL SAN RAFFAELE DI MILANO

Devolvi il tuo 5 per mille all’Ospedale San Raffaele: perché al centro della Ricerca ci sei TU. CODICE FISCALE: 07636600962, nel riquadro RICERCA SANITARIA. Non c’è cura, senza ricerca. Non c’è ricerca, senza il tuo 5xmille. Scopri come su http://www.5xmille.org

[http://www.5xmille.org/img/731x129.jpg]
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: stage top incubator

In reply to this post by Bozzo Luigi
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

You should also look at the Pathology devices incubator.  We have been using them on two microscopes for many years and they have worked very well.  We tried the Tokai Hit many years ago and chose the Pathology Devices instead.  No experience with the Oko, but we use their microscope cages.  Dave

Dr. David Knecht
Professor of Molecular and Cell Biology
Core Microscopy Facility Director
University of Connecticut
91 N. Eagleville Rd.
Storrs, CT 06269
860-486-2200

On Aug 23, 2016, at 7:56 AM, Bozzo Luigi <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all,

In our facility, we are looking to purchase a stage top incubator in order to do short term live cell imaging (5-30 min). We found two models that look compact (similar size of a wellplate) and user friendly: the Okolab H301-MINI and the Tokai Hit WELSX.

Does any anyone has positive or negative experience with one of those?


Any feedback to help us make a decision will be highly appreciated.

Thanks in advance,

Luigi


--
Luigi Bozzo, PhD
Bioimaging and Optics Platform (PT-BIOP)
Ecole Polytechnique Fédérale de Lausanne (EPFL)
Faculty of Life Sciences
Station 15, AI 0140
CH-1015 Lausanne

Phone: +41 21 693 9629
http://biop.epfl.ch/