transmited ligth and LSM

classic Classic list List threaded Threaded
♦
♦
13 messages Options
Maria Jimena Ortega Maria Jimena Ortega
Reply | Threaded
Open this post in threaded view
♦
♦
|

transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
 
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
 
Thanks a lot
 
Maria
Julio Vazquez Julio Vazquez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria, 

I think transmitted light images (including DIC and Phase) on a confocal will look pretty much the same as they would on a normal microscope.  The light path is reversed (illumination coming from the objective side, going though the sample, collected by the condenser, and detected on a PMT behind the condenser). These images will not be confocal.  On our LSM, DIC is actually very nice, better than on some of our other scopes, but I think it has to do with different implementations by different vendors.  We do not do phase on our confocals, so I can't comment much on that. I believe one can implement phase on a confocal, but I am not sure this is very widespread. Phase objectives have lower Numerical Apertures than their standard counterparts, and therefore are not ideal for fluorescence. DIC is more common, does not interfere with fluorescence imaging, and can substitute for Phase in most situations, except some special applications.

One important issue is that you need to have very stable lasers, otherwise you may see streaks or shading in your image. We improve our transmitted light images by averaging several frames.

Another issue is that confocals are generally used for the imaging of thick samples, while DIC and Phase work best with thin samples (such as cells). Transmitted light images of thick samples will generally not look very good, with Phase being the most sensitive probably. When imaging live cells and such on a confocal, however, you can have very nice DIC images in addition to your fluorescent channels.

Phase and DIC all use differences in refractive index to enhance contrast. This means that it works better with some samples (such as live cells in medium, where there are nice RI gradients between water and cell membranes, for instance), and not too well for others (such as fixed cells mounted in glycerol or other medium, where RI is very homogeneous).

What you can also do on a confocal, is reflected light, which uses the same confocal light path used for fluorescence (except that you detect the same wavelength band used for illumination). Works especially well for surfaces.

You can find more on this and related topics here:


and they have image galleries on various themes, such as DIC:




--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


==


On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
 
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
 
Thanks a lot
 
Maria

Nowell, Cameron Nowell, Cameron
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi Maria,
             We have two confocals in our facility that have DIC on them (an Olympus FV1000 and a Leica SP5). Both produce beuatiful DIC images (providing the prisims, filters etc are aligned properly). To be able to do transmitted light on a confocal it will need to have a transmitted light detector on it, these are usually seperate to the PMTs used for fluorecent imaging.
 
Also DIC will only be in focus in one plane of your image, you will not be able to get any confocal effect with DIC imaging.
 
I do not think phase would be able to work on a confocal due to the raster scanning pattern of the laser. Phase relies on difraction of a solid light beam.
 
Hope this helps. I can send you exapmle images of DIC captured on our systems if you want.
 
 
Cheers
 
 
Cam
 
 

Cameron Nowell 
Microscopy Research and Imaging Facility 
Cell Cycle and Development
Peter MacCallum Cancer Centre
7 St Andrews Place
East Melbourne, 3002
Victoria AUSTRALIA

Phone: +61396563759
Fax: +61396561411
Mobile: +61422882700



From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Julio Vazquez
Sent: Wednesday, 31 October 2007 5:03 AM
To: [hidden email]
Subject: Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria, 

I think transmitted light images (including DIC and Phase) on a confocal will look pretty much the same as they would on a normal microscope.  The light path is reversed (illumination coming from the objective side, going though the sample, collected by the condenser, and detected on a PMT behind the condenser). These images will not be confocal.  On our LSM, DIC is actually very nice, better than on some of our other scopes, but I think it has to do with different implementations by different vendors.  We do not do phase on our confocals, so I can't comment much on that. I believe one can implement phase on a confocal, but I am not sure this is very widespread. Phase objectives have lower Numerical Apertures than their standard counterparts, and therefore are not ideal for fluorescence. DIC is more common, does not interfere with fluorescence imaging, and can substitute for Phase in most situations, except some special applications.

One important issue is that you need to have very stable lasers, otherwise you may see streaks or shading in your image. We improve our transmitted light images by averaging several frames.

Another issue is that confocals are generally used for the imaging of thick samples, while DIC and Phase work best with thin samples (such as cells). Transmitted light images of thick samples will generally not look very good, with Phase being the most sensitive probably. When imaging live cells and such on a confocal, however, you can have very nice DIC images in addition to your fluorescent channels.

Phase and DIC all use differences in refractive index to enhance contrast. This means that it works better with some samples (such as live cells in medium, where there are nice RI gradients between water and cell membranes, for instance), and not too well for others (such as fixed cells mounted in glycerol or other medium, where RI is very homogeneous).

What you can also do on a confocal, is reflected light, which uses the same confocal light path used for fluorescence (except that you detect the same wavelength band used for illumination). Works especially well for surfaces.

You can find more on this and related topics here:


and they have image galleries on various themes, such as DIC:




--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


==


On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
 
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
 
Thanks a lot
 
Maria

This email (including any attachments) may contain
confidential and/or legally privileged information and is
intended only to be read or used by the addressee. If you
are not the intended addressee, any use, distribution,
disclosure or copying of this email is strictly
prohibited.
Confidentiality and legal privilege attached to this email
(including any attachments) are not waived or lost by
reason of its mistaken delivery to you.
If you have received this email in error, please delete it
and notify us immediately by telephone or email. Peter
MacCallum Cancer Centre provides no guarantee that this
transmission is free of virus or that it has not been
intercepted or altered and will not be liable for any delay
in its receipt.

George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmitted light and LSM

In reply to this post by Maria Jimena Ortega
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Maria,

The transmitted light images will look similar to what you see in the eyepiece. You should adjust the microscope for Koehler illumination. If your microscope has great DIC, like the current Leica SP5's, you will have great DIC on the computer, if your microscope has mediocre DIC, like my Axiovert 200M, you will have mediocre DIC. Phase contrast lenses are generally not as good in fluorescence as an  objective lens that does not have a phase ring - also the latter often have higher numerical aperture (brightness being proportional to NA to the 4th power).

You will probably need very low gain (I usually end up in the 200-260 V range on our LSM 510, and move the offset slider to make the scene darker. The transmitted light channel is also a handy way to see how stable - or not - your laser lines, AOTF, etc, are. Intensity tracks nicely with fluorescence specimens such as a Chroma plastic fluorescence slide, but without the possible issues of focus drift. Keep the PMT gain low!

Another valuable method is to use a reflected light (another term is "backscatter") channel. Set up a laser line to a PMT with appropriate dichroics and no emission filter (or a filter that transmits that line). You can:
* measure the coverglass thickness (need to account for its refractive index)
* coverglass-mounting medium interface location
* interference reflection contrast microscopy (IRM, aka RICM) image of cell-substratum interface - black spots are focal adhesion sites (see PubMed's Verschueren H.  1985 for details and related articles).
* (sometimes) reflections of plasma, organelle and nuclear membranes
* collagen and other fibers
* colloidal gold (I am looking forward to seeing how well it works with Nanoprobes's EnzMet products)
* medium-slide interface
Different laser lines will have different reflection patterns off the coverglass etc - if not wanted, they can sometimes be subtracted or otherwise digitally removed.


George
p.s. not transmitted light image, but I encourage everyone to check out my Halloween image http://www.nature.com/nprot/index.html Trick or treat! For those of you not familiar with Halloween, see http://en.wikipedia.org/wiki/Halloween though frankly the holiday seems to be an excuse to sell more candy. Mexican culture has much more interesting holiday - see  http://en.wikipedia.org/wiki/Day_of_the_Dead
I also encourage everyone to submit images to the website - they used three of mine as images of the week in the past couple of months, so they may be pretty desperate. Tip: Submit the same size image as on the site. Payoff; besides 7 days of fame, you can add a line item to your C.V.


At 08:26 AM 10/30/2007, you wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
 
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
 
Thanks a lot
 
Maria




 

George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility)


George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmitted light and LSM

In reply to this post by Julio Vazquez
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Julio's comments reminded me to mention that in transmitted light, a thin object will essentially disappear when in best focus. See Zernike 1955 Science 121: 345-349 ( http://www.sciencemag.org/cgi/reprint/121/3141/345) for why and what it led to.


At 02:02 PM 10/30/2007, you wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria,

I think transmitted light images (including DIC and Phase) on a confocal will look pretty much the same as they would on a normal microscope.  The light path is reversed (illumination coming from the objective side, going though the sample, collected by the condenser, and detected on a PMT behind the condenser). These images will not be confocal.  On our LSM, DIC is actually very nice, better than on some of our other scopes, but I think it has to do with different implementations by different vendors.  We do not do phase on our confocals, so I can't comment much on that. I believe one can implement phase on a confocal, but I am not sure this is very widespread. Phase objectives have lower Numerical Apertures than their standard counterparts, and therefore are not ideal for fluorescence. DIC is more common, does not interfere with fluorescence imaging, and can substitute for Phase in most situations, except some special applications.

One important issue is that you need to have very stable lasers, otherwise you may see streaks or shading in your image. We improve our transmitted light images by averaging several frames.

Another issue is that confocals are generally used for the imaging of thick samples, while DIC and Phase work best with thin samples (such as cells). Transmitted light images of thick samples will generally not look very good, with Phase being the most sensitive probably. When imaging live cells and such on a confocal, however, you can have very nice DIC images in addition to your fluorescent channels.

Phase and DIC all use differences in refractive index to enhance contrast. This means that it works better with some samples (such as live cells in medium, where there are nice RI gradients between water and cell membranes, for instance), and not too well for others (such as fixed cells mounted in glycerol or other medium, where RI is very homogeneous).

What you can also do on a confocal, is reflected light, which uses the same confocal light path used for fluorescence (except that you detect the same wavelength band used for illumination). Works especially well for surfaces.

You can find more on this and related topics here:

http://micro.magnet.fsu.edu/primer/techniques/index.html

and they have image galleries on various themes, such as DIC:

http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/index.html



--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024

http://www.fhcrc.org/

==


On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
 
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
 
Thanks a lot
 
Maria




 

George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility)


xavier Sanjuan xavier Sanjuan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

In reply to this post by Julio Vazquez
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Julio,
 
We can do both phase contrast and DIC in our Leica TCS SP2 system with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious about what exactly do you mean with "DIC does not interfere with fluorescence". One of the polarizers used for DIC is set between the sample and the detector, so it is somehow interfering the fluorescence lightpath. At least in our system, using the same laser and PMT settings, a fluorescence image got with that polarizer in will be darker than with the polarizer out. That's the reason why when somebody wants to get DIC and fluorescence images of a sample, we let that polarizer out of the system, getting still a good "pseudo" DIC image (just empirical, I can't explain why still a DIC image is formed without one of the filters needed to do it)).
 
Bst regards,
 
Xavi.

___________________________________

Xavier Sanjuan
Servei de Microscòpia Confocal
Departament de Ciències Experimentals i de la Salut
Universitat Pompeu Fabra
Parc de Recerca Biomèdica de Barcelona

Doctor Aiguader, 88 - Sala 309
08003 Barcelona - Spain

Tel.:  + 34 93 316 08 64
Fax: + 34 93 316 09 01
E-mail:
[hidden email]
Web: http://www.upf.edu/cexs/sct

 
-----Mensaje original-----
De: Confocal Microscopy List [mailto:[hidden email]]En nombre de Julio Vazquez
Enviado el: martes, 30 de octubre de 2007 19:03
Para: [hidden email]
Asunto: Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria,

I think transmitted light images (including DIC and Phase) on a confocal will look pretty much the same as they would on a normal microscope. The light path is reversed (illumination coming from the objective side, going though the sample, collected by the condenser, and detected on a PMT behind the condenser). These images will not be confocal. On our LSM, DIC is actually very nice, better than on some of our other scopes, but I think it has to do with different implementations by different vendors. We do not do phase on our confocals, so I can't comment much on that. I believe one can implement phase on a confocal, but I am not sure this is very widespread. Phase objectives have lower Numerical Apertures than their standard counterparts, and therefore are not ideal for fluorescence. DIC is more common, does not interfere with fluorescence imaging, and can substitute for Phase in most situations, except some special applications.

One important issue is that you need to have very stable lasers, otherwise you may see streaks or shading in your image. We improve our transmitted light images by averaging several frames.

Another issue is that confocals are generally used for the imaging of thick samples, while DIC and Phase work best with thin samples (such as cells). Transmitted light images of thick samples will generally not look very good, with Phase being the most sensitive probably. When imaging live cells and such on a confocal, however, you can have very nice DIC images in addition to your fluorescent channels.

Phase and DIC all use differences in refractive index to enhance contrast. This means that it works better with some samples (such as live cells in medium, where there are nice RI gradients between water and cell membranes, for instance), and not too well for others (such as fixed cells mounted in glycerol or other medium, where RI is very homogeneous).

What you can also do on a confocal, is reflected light, which uses the same confocal light path used for fluorescence (except that you detect the same wavelength band used for illumination). Works especially well for surfaces.

You can find more on this and related topics here:


and they have image galleries on various themes, such as DIC:




--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


==


On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
Thanks a lot
Maria

Stefan Terjung Stefan Terjung
Reply | Threaded
Open this post in threaded view
♦
♦
|

AW: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear Xavier,
 
as the lasers are already polarized you don´t need the polarizer (between scanhead and objective lens) in when taking DIC images with your SP2 (but you still need the polarizer between condensor and transmission detector). So you already do the right thing by leaving it out, it´s not "pseudo" DIC. And I guess this is what Julio meant with "DIC does not interfere with fluorescence" (Ok still the DIC prism is in, but this doesn´t decrease the detected fluorescence intensity significantly).
 
regards,
 
Stefan
 

------------------------
Dr. Stefan Terjung
Advanced Light Microscopy Facility
EMBL Heidelberg
Meyerhofstr.1
D - 69117 Heidelberg
Phone +49(0)6221 387-8467



Von: Confocal Microscopy List [mailto:[hidden email]] Im Auftrag von xavier Sanjuan
Gesendet: Mittwoch, 31. Oktober 2007 11:03
An: [hidden email]
Betreff: Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Julio,
 
We can do both phase contrast and DIC in our Leica TCS SP2 system with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious about what exactly do you mean with "DIC does not interfere with fluorescence". One of the polarizers used for DIC is set between the sample and the detector, so it is somehow interfering the fluorescence lightpath. At least in our system, using the same laser and PMT settings, a fluorescence image got with that polarizer in will be darker than with the polarizer out. That's the reason why when somebody wants to get DIC and fluorescence images of a sample, we let that polarizer out of the system, getting still a good "pseudo" DIC image (just empirical, I can't explain why still a DIC image is formed without one of the filters needed to do it)).
 
Bst regards,
 
Xavi.

___________________________________

Xavier Sanjuan
Servei de Microscòpia Confocal
Departament de Ciències Experimentals i de la Salut
Universitat Pompeu Fabra
Parc de Recerca Biomèdica de Barcelona

Doctor Aiguader, 88 - Sala 309
08003 Barcelona - Spain

Tel.:  + 34 93 316 08 64
Fax: + 34 93 316 09 01
E-mail:
[hidden email]
Web: http://www.upf.edu/cexs/sct

 
-----Mensaje original-----
De: Confocal Microscopy List [mailto:[hidden email]]En nombre de Julio Vazquez
Enviado el: martes, 30 de octubre de 2007 19:03
Para: [hidden email]
Asunto: Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria,

I think transmitted light images (including DIC and Phase) on a confocal will look pretty much the same as they would on a normal microscope. The light path is reversed (illumination coming from the objective side, going though the sample, collected by the condenser, and detected on a PMT behind the condenser). These images will not be confocal. On our LSM, DIC is actually very nice, better than on some of our other scopes, but I think it has to do with different implementations by different vendors. We do not do phase on our confocals, so I can't comment much on that. I believe one can implement phase on a confocal, but I am not sure this is very widespread. Phase objectives have lower Numerical Apertures than their standard counterparts, and therefore are not ideal for fluorescence. DIC is more common, does not interfere with fluorescence imaging, and can substitute for Phase in most situations, except some special applications.

One important issue is that you need to have very stable lasers, otherwise you may see streaks or shading in your image. We improve our transmitted light images by averaging several frames.

Another issue is that confocals are generally used for the imaging of thick samples, while DIC and Phase work best with thin samples (such as cells). Transmitted light images of thick samples will generally not look very good, with Phase being the most sensitive probably. When imaging live cells and such on a confocal, however, you can have very nice DIC images in addition to your fluorescent channels.

Phase and DIC all use differences in refractive index to enhance contrast. This means that it works better with some samples (such as live cells in medium, where there are nice RI gradients between water and cell membranes, for instance), and not too well for others (such as fixed cells mounted in glycerol or other medium, where RI is very homogeneous).

What you can also do on a confocal, is reflected light, which uses the same confocal light path used for fluorescence (except that you detect the same wavelength band used for illumination). Works especially well for surfaces.

You can find more on this and related topics here:


and they have image galleries on various themes, such as DIC:




--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA 98109-1024


==


On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello everybody!
I`d like to ask you what I should expect from my LSM when trying to get transmited images...
Is somewhere in the web any image gallery you could recommend?
How much will DIC, Ph, etc improve?
Thanks a lot
Maria

Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

In reply to this post by xavier Sanjuan
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: transmited ligth and LSM We have trouble doing this on our SP2, because the polarity of the lasers has never been stable, drives us crazy.  So DIC images oscillate from light to dark and back again – within a single image.  The engineers have never been able to fix it.

The reason you can do DIC with only one polariser in the system is because the laser itself is polarised.  But if you have a Wollaston prism between scan head and specimen when collecting a fluorescence image, it will be degraded, because you’ll be able to detect the prism interacting with the beam – the image is scattered, reducing resolution.

Cheers,
rosemary

Dr Rosemary White               [hidden email]
CSIRO Plant Industry            ph.     02-6246 5475
GPO Box 1600                        fax.     02-6246 5334
Canberra, ACT 2601            
Australia


On 31/10/07 9:02 PM, "xavier Sanjuan" <[hidden email]> wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Julio,

We can do both phase contrast and DIC in our Leica TCS SP2 system with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious about what exactly do you mean with "DIC does not interfere with fluorescence". One of the polarizers used for DIC is set between the sample and the detector, so it is somehow interfering the fluorescence lightpath. At least in our system, using the same laser and PMT settings, a fluorescence image got with that polarizer in will be darker than with the polarizer out. That's the reason why when somebody wants to get DIC and fluorescence images of a sample, we let that polarizer out of the system, getting still a good "pseudo" DIC image (just empirical, I can't explain why still a DIC image is formed without one of the filters needed to do it)).

Bst regards,

Xavi.
___________________________________

Xavier Sanjuan
Servei de Microscòpia Confocal
Departament de Ciències Experimentals i de la Salut
Universitat Pompeu Fabra
Parc de Recerca Biomèdica de Barcelona
Doctor Aiguader, 88 - Sala 309
08003 Barcelona - Spain

Tel.:  + 34 93 316 08 64
Fax: + 34 93 316 09 01
E-mail: [hidden email]
[hidden email]
Web: http://www.upf.edu/cexs/sct <http://www.upf.edu/cexs/sct>

 

-----Mensaje original-----
De: Confocal Microscopy List  [[hidden email]En nombre de Julio  Vazquez
Enviado el: martes, 30 de octubre de 2007  19:03
Para: [hidden email]
Asunto: Re:  transmited ligth and LSM

Search the CONFOCAL archive at  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria,
 

 
I think transmitted light images (including DIC and Phase) on a confocal  will look pretty much the same as they would on a normal microscope. The light  path is reversed (illumination coming from the objective side, going though  the sample, collected by the condenser, and detected on a PMT behind the  condenser). These images will not be confocal. On our LSM, DIC is actually  very nice, better than on some of our other scopes, but I think it has to do  with different implementations by different vendors. We do not do phase on our  confocals, so I can't comment much on that. I believe one can implement phase  on a confocal, but I am not sure this is very widespread. Phase objectives  have lower Numerical Apertures than their standard counterparts, and therefore  are not ideal for fluorescence. DIC is more common, does not interfere with  fluorescence imaging, and can substitute for Phase in most situations, except  some special applications.
 

 
One important issue is that you need to have very stable lasers,  otherwise you may see streaks or shading in your image. We improve our  transmitted light images by averaging several frames.
 

 
Another issue is that confocals are generally used for the imaging of  thick samples, while DIC and Phase work best with thin samples (such as  cells). Transmitted light images of thick samples will generally not look very  good, with Phase being the most sensitive probably. When imaging live cells  and such on a confocal, however, you can have very nice DIC images in addition  to your fluorescent channels.
 

 
Phase and DIC all use differences in refractive index to enhance  contrast. This means that it works better with some samples (such as live  cells in medium, where there are nice RI gradients between water and cell  membranes, for instance), and not too well for others (such as fixed cells  mounted in glycerol or other medium, where RI is very homogeneous).
 

 
What you can also do on a confocal, is reflected light, which uses the  same confocal light path used for fluorescence (except that you detect the  same wavelength band used for illumination). Works especially well for  surfaces.
 

 
You can find more on this and related topics here:
 

 
http://micro.magnet.fsu.edu/primer/techniques/index.html
 

 
and they have image galleries on various themes, such as DIC:
 

 
http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/index.html
 

 

 

 
 
--
 
Julio Vazquez
 
Fred Hutchinson Cancer Research Center
 
Seattle, WA 98109-1024
 

 
http://www.fhcrc.org/
 

 
==


 
 
On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

 
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  
Hello everybody!
 
 
I`d like to ask you what I should expect from my LSM when  trying to get transmited images...
 
Is somewhere in the web any image gallery you could  recommend?
 
How much will DIC, Ph, etc improve?
 
 
Thanks a lot
 
 
Maria



Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal We also have problems with DIC on our SP2 which have never been fixed. In our case it is sometimes rings of light dark, and sometimes serious gradients across the field.   Dave

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


On Oct 31, 2007, at 6:52 AM, Rosemary White wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal We have trouble doing this on our SP2, because the polarity of the lasers has never been stable, drives us crazy.  So DIC images oscillate from light to dark and back again – within a single image.  The engineers have never been able to fix it.

The reason you can do DIC with only one polariser in the system is because the laser itself is polarised.  But if you have a Wollaston prism between scan head and specimen when collecting a fluorescence image, it will be degraded, because you’ll be able to detect the prism interacting with the beam – the image is scattered, reducing resolution.

Cheers,
rosemary

Dr Rosemary White               [hidden email]
CSIRO Plant Industry            ph.     02-6246 5475
GPO Box 1600                        fax.     02-6246 5334
Canberra, ACT 2601            
Australia


On 31/10/07 9:02 PM, "xavier Sanjuan" <[hidden email]> wrote:

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Julio,

We can do both phase contrast and DIC in our Leica TCS SP2 system with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious about what exactly do you mean with "DIC does not interfere with fluorescence". One of the polarizers used for DIC is set between the sample and the detector, so it is somehow interfering the fluorescence lightpath. At least in our system, using the same laser and PMT settings, a fluorescence image got with that polarizer in will be darker than with the polarizer out. That's the reason why when somebody wants to get DIC and fluorescence images of a sample, we let that polarizer out of the system, getting still a good "pseudo" DIC image (just empirical, I can't explain why still a DIC image is formed without one of the filters needed to do it)).

Bst regards,

Xavi.
___________________________________

Xavier Sanjuan
Servei de Microscòpia Confocal
Departament de Ciències Experimentals i de la Salut
Universitat Pompeu Fabra
Parc de Recerca Biomèdica de Barcelona
Doctor Aiguader, 88 - Sala 309
08003 Barcelona - Spain

Tel.:  + 34 93 316 08 64
Fax: + 34 93 316 09 01
E-mail: [hidden email]
[hidden email]
Web: http://www.upf.edu/cexs/sct <http://www.upf.edu/cexs/sct>

 

-----Mensaje original-----
De: Confocal Microscopy List  [[hidden email]En nombre de Julio  Vazquez
Enviado el: martes, 30 de octubre de 2007  19:03
Para: [hidden email]
Asunto: Re:  transmited ligth and LSM

Search the CONFOCAL archive at  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
Hello Maria,
 

 
I think transmitted light images (including DIC and Phase) on a confocal  will look pretty much the same as they would on a normal microscope. The light  path is reversed (illumination coming from the objective side, going though  the sample, collected by the condenser, and detected on a PMT behind the  condenser). These images will not be confocal. On our LSM, DIC is actually  very nice, better than on some of our other scopes, but I think it has to do  with different implementations by different vendors. We do not do phase on our  confocals, so I can't comment much on that. I believe one can implement phase  on a confocal, but I am not sure this is very widespread. Phase objectives  have lower Numerical Apertures than their standard counterparts, and therefore  are not ideal for fluorescence. DIC is more common, does not interfere with  fluorescence imaging, and can substitute for Phase in most situations, except  some special applications.
 

 
One important issue is that you need to have very stable lasers,  otherwise you may see streaks or shading in your image. We improve our  transmitted light images by averaging several frames.
 

 
Another issue is that confocals are generally used for the imaging of  thick samples, while DIC and Phase work best with thin samples (such as  cells). Transmitted light images of thick samples will generally not look very  good, with Phase being the most sensitive probably. When imaging live cells  and such on a confocal, however, you can have very nice DIC images in addition  to your fluorescent channels.
 

 
Phase and DIC all use differences in refractive index to enhance  contrast. This means that it works better with some samples (such as live  cells in medium, where there are nice RI gradients between water and cell  membranes, for instance), and not too well for others (such as fixed cells  mounted in glycerol or other medium, where RI is very homogeneous).
 

 
What you can also do on a confocal, is reflected light, which uses the  same confocal light path used for fluorescence (except that you detect the  same wavelength band used for illumination). Works especially well for  surfaces.
 

 
You can find more on this and related topics here:
 

 
http://micro.magnet.fsu.edu/primer/techniques/index.html
 

 
and they have image galleries on various themes, such as DIC:
 

 
http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/index.html
 

 

 

 
 
--
 
Julio Vazquez
 
Fred Hutchinson Cancer Research Center
 
Seattle, WA 98109-1024
 

 
http://www.fhcrc.org/
 

 
==


 
 
On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:

 
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  
Hello everybody!
 
 
I`d like to ask you what I should expect from my LSM when  trying to get transmited images...
 
Is somewhere in the web any image gallery you could  recommend?
 
How much will DIC, Ph, etc improve?
 
 
Thanks a lot
 
 
Maria




JOEL B. SHEFFIELD JOEL B. SHEFFIELD
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi David,

We have no trouble with our SP2, except with a LWD 40 dry objective.  
I think the issue of gradients is often one of a mismatch between the
Wollaston prism and the objective.

Also, check the settings of your condenser aperture.  On our system,
that aperture has a profound effect on the DIC images.

I find that there is greater fluctuation with our Argon laser than
with the HeNe green and red lasers.  Since we often use sequential
scan, I do the DIC image with the green laser source.

Joel


-------------- Original message ---------------
Date sent: Wed, 31 Oct 2007 08:26:44 -0400
Send reply to: Confocal Microscopy List
<[hidden email]>
From: David Knecht <[hidden email]>
Subject: Re: transmited ligth and LSM
To: [hidden email]

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-
bin/wa?S1=confocal We also have problems with DIC on our SP2 which
have never been fixed. In our case it is sometimes rings of light
dark, and sometimes serious gradients across the field.  Dave
Dr. David Knecht  
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)



On Oct 31, 2007, at 6:52 AM, Rosemary White wrote:





Joel B. Sheffield, Ph.D
Department of Biology
Temple University
Philadelphia, PA 19122
Voice: 215 204 8839
e-mail: [hidden email]
URL:  http://astro.temple.edu/~jbs
Judy Trogadis Judy Trogadis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

In reply to this post by xavier Sanjuan
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

You don't need a polarizer with laser light, so you should remove it when capturing images. It is there only for observation prior to image capture.

Judy

Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]


>>> xavier Sanjuan <[hidden email]> 10/31/2007 6:02 AM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

Hello Julio,

We can do both phase contrast and DIC in our Leica TCS SP2 system with 40x
1.25 and 63x 1.32 oil immersion objectives. I'm curious about what exactly
do you mean with "DIC does not interfere with fluorescence". One of the
polarizers used for DIC is set between the sample and the detector, so it is
somehow interfering the fluorescence lightpath. At least in our system,
using the same laser and PMT settings, a fluorescence image got with that
polarizer in will be darker than with the polarizer out. That's the reason
why when somebody wants to get DIC and fluorescence images of a sample, we
let that polarizer out of the system, getting still a good "pseudo" DIC
image (just empirical, I can't explain why still a DIC image is formed
without one of the filters needed to do it)).

Bst regards,

Xavi.
___________________________________

Xavier Sanjuan
Servei de Microscopia Confocal
Departament de Ciencies Experimentals i de la Salut
Universitat Pompeu Fabra
Parc de Recerca Biomedica de Barcelona
Doctor Aiguader, 88 - Sala 309
08003 Barcelona - Spain

Tel.:  + 34 93 316 08 64
Fax: + 34 93 316 09 01
E-mail: [hidden email]
Web: http://www.upf.edu/cexs/sct 


  -----Mensaje original-----
  De: Confocal Microscopy List [mailto:[hidden email]]En
nombre de Julio Vazquez
  Enviado el: martes, 30 de octubre de 2007 19:03
  Para: [hidden email]
  Asunto: Re: transmited ligth and LSM


  Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
  Hello Maria,


  I think transmitted light images (including DIC and Phase) on a confocal
will look pretty much the same as they would on a normal microscope. The
light path is reversed (illumination coming from the objective side, going
though the sample, collected by the condenser, and detected on a PMT behind
the condenser). These images will not be confocal. On our LSM, DIC is
actually very nice, better than on some of our other scopes, but I think it
has to do with different implementations by different vendors. We do not do
phase on our confocals, so I can't comment much on that. I believe one can
implement phase on a confocal, but I am not sure this is very widespread.
Phase objectives have lower Numerical Apertures than their standard
counterparts, and therefore are not ideal for fluorescence. DIC is more
common, does not interfere with fluorescence imaging, and can substitute for
Phase in most situations, except some special applications.


  One important issue is that you need to have very stable lasers, otherwise
you may see streaks or shading in your image. We improve our transmitted
light images by averaging several frames.


  Another issue is that confocals are generally used for the imaging of
thick samples, while DIC and Phase work best with thin samples (such as
cells). Transmitted light images of thick samples will generally not look
very good, with Phase being the most sensitive probably. When imaging live
cells and such on a confocal, however, you can have very nice DIC images in
addition to your fluorescent channels.


  Phase and DIC all use differences in refractive index to enhance contrast.
This means that it works better with some samples (such as live cells in
medium, where there are nice RI gradients between water and cell membranes,
for instance), and not too well for others (such as fixed cells mounted in
glycerol or other medium, where RI is very homogeneous).


  What you can also do on a confocal, is reflected light, which uses the
same confocal light path used for fluorescence (except that you detect the
same wavelength band used for illumination). Works especially well for
surfaces.


  You can find more on this and related topics here:


  http://micro.magnet.fsu.edu/primer/techniques/index.html 


  and they have image galleries on various themes, such as DIC:


  http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/index.html 






  --
  Julio Vazquez
  Fred Hutchinson Cancer Research Center
  Seattle, WA 98109-1024


  http://www.fhcrc.org/ 


  ==




  On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:


    Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
    Hello everybody!
    I`d like to ask you what I should expect from my LSM when trying to get
transmited images...
    Is somewhere in the web any image gallery you could recommend?
    How much will DIC, Ph, etc improve?
    Thanks a lot
    Maria
Judy Trogadis Judy Trogadis
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: transmited ligth and LSM

In reply to this post by Knecht, David
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have just gone through this same problem with our own SP2 system. The
rings are there during brightfield observation at 10x and according to
the Leica people, cannot be eliminated. If you zoom even a little, the
rings will be out of the field of view. We decided not to answer
questions at 10x brightfield ;-)

The gradient issue could be due to misalignment of the fiber optic
detector.

Judy

Judy Trogadis
Bio-Imaging Coordinator
St. Michael's Hospital, 7Queen
30 Bond St.
Toronto, ON M5B 1W8, Canada
ph:  416-864-6060  x6337
pager: 416-685-9219
fax: 416-864-6043
[hidden email]


>>> David Knecht <[hidden email]> 10/31/2007 8:26 AM >>>
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

We also have problems with DIC on our SP2 which have never been  
fixed. In our case it is sometimes rings of light dark, and sometimes

serious gradients across the field.   Dave

Dr. David Knecht
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


On Oct 31, 2007, at 6:52 AM, Rosemary White wrote:

> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-

> bin/wa?S1=confocal We have trouble doing this on our SP2, because  
> the polarity of the lasers has never been stable, drives us crazy.  

> So DIC images oscillate from light to dark and back again * within

> a single image.  The engineers have never been able to fix it.
>
> The reason you can do DIC with only one polariser in the system is  
> because the laser itself is polarised.  But if you have a Wollaston

> prism between scan head and specimen when collecting a fluorescence

> image, it will be degraded, because you*ll be able to detect the  
> prism interacting with the beam * the image is scattered, reducing

> resolution.
>
> Cheers,
> rosemary
>
> Dr Rosemary White               [hidden email]
> CSIRO Plant Industry            ph.     02-6246 5475
> GPO Box 1600                        fax.     02-6246 5334
> Canberra, ACT 2601
> Australia
>
>
> On 31/10/07 9:02 PM, "xavier Sanjuan" <[hidden email]>
wrote:

>
>> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/ 
>> cgi-bin/wa?S1=confocal
>> Hello Julio,
>>
>> We can do both phase contrast and DIC in our Leica TCS SP2 system  
>> with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious  
>> about what exactly do you mean with "DIC does not interfere with  
>> fluorescence". One of the polarizers used for DIC is set between  
>> the sample and the detector, so it is somehow interfering the  
>> fluorescence lightpath. At least in our system, using the same  
>> laser and PMT settings, a fluorescence image got with that  
>> polarizer in will be darker than with the polarizer out. That's  
>> the reason why when somebody wants to get DIC and fluorescence  
>> images of a sample, we let that polarizer out of the system,  
>> getting still a good "pseudo" DIC image (just empirical, I can't  
>> explain why still a DIC image is formed without one of the filters

>> needed to do it)).
>>
>> Bst regards,
>>
>> Xavi.
>> ___________________________________
>>
>> Xavier Sanjuan
>> Servei de Microscòpia Confocal
>> Departament de Ciències Experimentals i de la Salut
>> Universitat Pompeu Fabra
>> Parc de Recerca Biomèdica de Barcelona
>> Doctor Aiguader, 88 - Sala 309
>> 08003 Barcelona - Spain
>>
>> Tel.:  + 34 93 316 08 64
>> Fax: + 34 93 316 09 01
>> E-mail: [hidden email] <mailto:[hidden email]>
>> Web: http://www.upf.edu/cexs/sct <http://www.upf.edu/cexs/sct>
>>
>>
>>>
>>> -----Mensaje original-----
>>> De: Confocal Microscopy List  
>>> [mailto:[hidden email]]En nombre de Julio  Vazquez
>>> Enviado el: martes, 30 de octubre de 2007  19:03
>>> Para: CON
[hidden email]

>>> Asunto: Re:  transmited ligth and LSM
>>>
>>> Search the CONFOCAL archive at  http://listserv.acsu.buffalo.edu/ 
>>> cgi-bin/wa?S1=confocal -
>>> Hello Maria,
>>>
>>>
>>>
>>> I think transmitted light images (including DIC and Phase) on a  
>>> confocal  will look pretty much the same as they would on a  
>>> normal microscope. The light  path is reversed (illumination  
>>> coming from the objective side, going though  the sample,  
>>> collected by the condenser, and detected on a PMT behind the  
>>> condenser). These images will not be confocal. On our LSM, DIC is

>>> actually  very nice, better than on some of our other scopes, but

>>> I think it has to do  with different implementations by different

>>> vendors. We do not do phase on our  confocals, so I can't comment

>>> much on that. I believe one can implement phase  on a confocal,  
>>> but I am not sure this is very widespread. Phase objectives  have

>>> lower Numerical Apertures than their standard counterparts, and  
>>> therefore  are not ideal for fluorescence. DIC is more common,  
>>> does not interfere with  fluorescence imaging, and can substitute

>>> for Phase in most situations, except  some special applications.
>>>
>>>
>>>
>>> One important issue is that you need to have very stable lasers,  

>>> otherwise you may see streaks or shading in your image. We  
>>> improve our  transmitted light images by averaging several frames.
>>>
>>>
>>>
>>> Another issue is that confocals are generally used for the  
>>> imaging of  thick samples, while DIC and Phase work best with  
>>> thin samples (such as  cells). Transmitted light images of thick  
>>> samples will generally not look very  good, with Phase being the  
>>> most sensitive probably. When imaging live cells  and such on a  
>>> confocal, however, you can have very nice DIC images in addition  

>>> to your fluorescent channels.
>>>
>>>
>>>
>>> Phase and DIC all use differences in refractive index to enhance  

>>> contrast. This means that it works better with some samples (such

>>> as live  cells in medium, where there are nice RI gradients  
>>> between water and cell  membranes, for instance), and not too  
>>> well for others (such as fixed cells  mounted in glycerol or  
>>> other medium, where RI is very homogeneous).
>>>
>>>
>>>
>>> What you can also do on a confocal, is reflected light, which  
>>> uses the  same confocal light path used for fluorescence (except  
>>> that you detect the  same wavelength band used for illumination).

>>> Works especially well for  surfaces.
>>>
>>>
>>>
>>> You can find more on this and related topics here:
>>>
>>>
>>>
>>> http://micro.magnet.fsu.edu/primer/techniques/index.html 
>>>
>>>
>>>
>>> and they have image galleries on various themes, such as DIC:
>>>
>>>
>>>
>>> http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/ 
>>> index.html
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> --
>>>
>>> Julio Vazquez
>>>
>>> Fred Hutchinson Cancer Research Center
>>>
>>> Seattle, WA 98109-1024
>>>
>>>
>>>
>>> http://www.fhcrc.org/ 
>>>
>>>
>>>
>>> ==
>>>
>>>
>>>
>>>
>>> On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:
>>>
>>>
>>>> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/ 
>>>> cgi-bin/wa?S1=confocal
>>>> Hello everybody!
>>>>
>>>>
>>>> I`d like to ask you what I should expect from my LSM when  
>>>> trying to get transmited images...
>>>>
>>>> Is somewhere in the web any image gallery you could  recommend?
>>>>
>>>> How much will DIC, Ph, etc improve?
>>>>
>>>>
>>>> Thanks a lot
>>>>
>>>>
>>>> Maria
>>>
>>
>
Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: AW: transmited ligth and LSM

In reply to this post by Stefan Terjung
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Stefan,

in fact the DIC prism does interfer significantly with the image, it has
quite some impact at resolution. This can be easily observed with small
beads, already 500nm ones will look quite scattered.

On a Leica DMI6000 the DIC prisms can be located in an additional
turrent, which means you can acquire DIC (with prism in) sequentially to
fluorescence (with prism out).

cheers,
Michael


Stefan Terjung wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Dear Xavier,
>  
> as the lasers are already polarized you don´t need the polarizer
> (between scanhead and objective lens) in when taking DIC images with
> your SP2 (but you still need the polarizer between condensor and
> transmission detector). So you already do the right thing by leaving it
> out, it´s not "pseudo" DIC. And I guess this is what Julio meant with
> "DIC does not interfere with fluorescence" (Ok still the DIC prism is
> in, but this doesn´t decrease the detected fluorescence intensity
> significantly).
>  
> regards,
>  
> Stefan
>  
>
> ------------------------
> Dr. Stefan Terjung
> Advanced Light Microscopy Facility
> EMBL Heidelberg
> Meyerhofstr.1
> D - 69117 Heidelberg
> Phone +49(0)6221 387-8467
>
>
>     ------------------------------------------------------------------------
>     *Von:* Confocal Microscopy List
>     [mailto:[hidden email]] *Im Auftrag von *xavier Sanjuan
>     *Gesendet:* Mittwoch, 31. Oktober 2007 11:03
>     *An:* [hidden email]
>     *Betreff:* Re: transmited ligth and LSM
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>     Hello Julio,
>      
>     We can do both phase contrast and DIC in our Leica TCS SP2 system
>     with 40x 1.25 and 63x 1.32 oil immersion objectives. I'm curious
>     about what exactly do you mean with "DIC does not interfere with
>     fluorescence". One of the polarizers used for DIC is set between the
>     sample and the detector, so it is somehow interfering the
>     fluorescence lightpath. At least in our system, using the same laser
>     and PMT settings, a fluorescence image got with that polarizer in
>     will be darker than with the polarizer out. That's the reason why
>     when somebody wants to get DIC and fluorescence images of a sample,
>     we let that polarizer out of the system, getting still a good
>     "pseudo" DIC image (just empirical, I can't explain why still a DIC
>     image is formed without one of the filters needed to do it)).
>      
>     Bst regards,
>      
>     Xavi.
>
>     /___________________________________/
>
>     ///Xavier Sanjuan//
>     //Servei de Microscòpia Confocal//
>     /Departament de Ciències Experimentals i de la Salut/
>     /Universitat Pompeu Fabra /
>     /Parc de Recerca Biomèdica de Barcelona//
>     /Doctor Aiguader, 88 - Sala 309
>     08003 Barcelona - Spain
>
>     Tel.:  + 34 93 316 08 64
>     Fax: + 34 93 316 09 01
>     E-mail: //[hidden email]/ <mailto:[hidden email]>
>     /Web: //http://www.upf.edu/cexs/sct/
>
>      
>
>         -----Mensaje original-----
>         *De:* Confocal Microscopy List
>         [mailto:[hidden email]]*En nombre de *Julio Vazquez
>         *Enviado el:* martes, 30 de octubre de 2007 19:03
>         *Para:* [hidden email]
>         *Asunto:* Re: transmited ligth and LSM
>
>         Search the CONFOCAL archive at
>         http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal -
>         Hello Maria,
>
>         I think transmitted light images (including DIC and Phase) on a
>         confocal will look pretty much the same as they would on a
>         normal microscope. The light path is reversed (illumination
>         coming from the objective side, going though the sample,
>         collected by the condenser, and detected on a PMT behind the
>         condenser). These images will not be confocal. On our LSM, DIC
>         is actually very nice, better than on some of our other scopes,
>         but I think it has to do with different implementations by
>         different vendors. We do not do phase on our confocals, so I
>         can't comment much on that. I believe one can implement phase on
>         a confocal, but I am not sure this is very widespread. Phase
>         objectives have lower Numerical Apertures than their standard
>         counterparts, and therefore are not ideal for fluorescence. DIC
>         is more common, does not interfere with fluorescence imaging,
>         and can substitute for Phase in most situations, except some
>         special applications.
>
>         One important issue is that you need to have very stable lasers,
>         otherwise you may see streaks or shading in your image. We
>         improve our transmitted light images by averaging several frames.
>
>         Another issue is that confocals are generally used for the
>         imaging of thick samples, while DIC and Phase work best with
>         thin samples (such as cells). Transmitted light images of thick
>         samples will generally not look very good, with Phase being the
>         most sensitive probably. When imaging live cells and such on a
>         confocal, however, you can have very nice DIC images in addition
>         to your fluorescent channels.
>
>         Phase and DIC all use differences in refractive index to enhance
>         contrast. This means that it works better with some samples
>         (such as live cells in medium, where there are nice RI gradients
>         between water and cell membranes, for instance), and not too
>         well for others (such as fixed cells mounted in glycerol or
>         other medium, where RI is very homogeneous).
>
>         What you can also do on a confocal, is reflected light, which
>         uses the same confocal light path used for fluorescence (except
>         that you detect the same wavelength band used for illumination).
>         Works especially well for surfaces.
>
>         You can find more on this and related topics here:
>
>         http://micro.magnet.fsu.edu/primer/techniques/index.html
>
>         and they have image galleries on various themes, such as DIC:
>
>         http://micro.magnet.fsu.edu/primer/techniques/dic/dicgallery/index.html
>
>
>
>         --
>         Julio Vazquez
>         Fred Hutchinson Cancer Research Center
>         Seattle, WA 98109-1024
>
>         http://www.fhcrc.org/
>
>         ==
>
>
>         On Oct 30, 2007, at 5:26 AM, Maria Jimena Ortega wrote:
>
>>         Search the CONFOCAL archive at
>>         http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>         Hello everybody!
>>         I`d like to ask you what I should expect from my LSM when
>>         trying to get transmited images...
>>         Is somewhere in the web any image gallery you could recommend?
>>         How much will DIC, Ph, etc improve?
>>         Thanks a lot
>>         Maria