Sylvie Le Guyader-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all We have a 60x water objective NA1.2 by Nikon that we would like to use to screen 96 and 384 well plates. Does anyone know of a water refill system that would fit on this objective? Or has anyone tried the Olympus or the Leica water refill device? Do they work well? Would any fit our objective? Thanks for your help! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet Sweden office: +46 (0)8 608 9240 mobile: +46 (0) 73 733 5008 |
![]() ![]() |
Boswell, Carl A - (cboswell) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The Olympus FV10i: http://www.olympusamerica.com/seg_section/product.asp?product=1064 does what you are looking for. Maybe you can get some info from them on their source of hardware/setup. c Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Sylvie Le Guyader" <[hidden email]> To: <[hidden email]> Sent: Friday, September 24, 2010 6:03 AM Subject: water refill device for WI objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all We have a 60x water objective NA1.2 by Nikon that we would like to use to screen 96 and 384 well plates. Does anyone know of a water refill system that would fit on this objective? Or has anyone tried the Olympus or the Leica water refill device? Do they work well? Would any fit our objective? Thanks for your help! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet Sweden office: +46 (0)8 608 9240 mobile: +46 (0) 73 733 5008 |
![]() ![]() |
Sylvie Le Guyader-2 |
![]() |
In reply to this post by Sylvie Le Guyader-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Tom Thanks for your suggestion. We use the Zeiss oil for water lens routinely and it works very well as long as one doesnt try to image a whole multiwell plate. Oil (normal or for water lenses) is too viscous for large areas. It collects at places and goes missing at others. We have tried this a lot without success. The only option we have found to combine higher throughput and high resolution is to use a water objective. It works really nicely with glass bottom plates but now we also acquire time lapses, so I guess the only way to go is to have a water refill system. Hence my query... Does anyone have any experience with this type of protocols? Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet Sweden office: +46 (0)8 608 9240 LCI room: +46 (0)8 608 9248 mobile: +46 (0) 73 733 5008 > -----Original Message----- > From: Donnelly, Tom [mailto:[hidden email]] > Sent: 24 September 2010 19:11 > To: '[hidden email]' > Subject: RE: water refill device for WI objective > > Sylvie; > > I have often thought the oil that Zeiss offers that has the same refractive index of > water might be a good solution here. Once the plate is coated ( probably do that > before starting scan) it would seem that little additional oil would be needed. > > I have no experience with this and have no idea if it is compatible with Nikon > objectives, but just a thought. > > Tom > > Tom Donnelly > Applied Precision, Inc. > 1040 12th Ave. N.W. Issaquah, WA 98027-8929 > (425)829-9414 (425)557-1055 fax > [hidden email] > http://www.appliedprecision.com > > > > -----Original Message----- > From: Sylvie Le Guyader [mailto:[hidden email]] > Sent: Friday, September 24, 2010 6:04 AM > To: [hidden email] > Subject: water refill device for WI objective > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all > > > > We have a 60x water objective NA1.2 by Nikon that we would like to use to > 96 and 384 well plates. Does anyone know of a water refill system that would fit on > this objective? Or has anyone tried the Olympus or the Leica water refill device? Do > they work well? Would any fit our objective? > > > > Thanks for your help! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Live Cell Imaging Unit > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Sweden > > office: +46 (0)8 608 9240 > > mobile: +46 (0) 73 733 5008 > > > > > This email message, together with any attachments, is for the sole use of > intended recipient(s) and is the confidential information of Applied Precision Inc. If > you are not the intended recipient, your review, use, disclosure, copying or > dissemination of this email message or its attachments, or the information > contained therein, is strictly prohibited. If you are not the intended recipient or if you > think this email was sent to you in error, please notify the sender by reply email and > delete this message and its attachments, as well as all copies, from your system. |
![]() ![]() |
Kurt Thorn |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I believe Sutter was working on a custom water pump system to refill water on an objective, but I'm not sure if it came together or not. I've also wondered if silanizing the bottom of the plate to make it more hydrophobic would give a more uniform oil distribution. Kurt On 9/27/2010 6:49 AM, Sylvie Le Guyader wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Tom > > Thanks for your suggestion. We use the Zeiss oil for water lens routinely > and it works very well as long as one doesn’t try to image a whole multiwell > plate. Oil (normal or for water lenses) is too viscous for large areas. It > collects at places and goes missing at others. We have tried this a lot > without success. The only option we have found to combine higher throughput > and high resolution is to use a water objective. It works really nicely with > glass bottom plates but now we also acquire time lapses, so I guess the only > way to go is to have a water refill system. Hence my query... Does anyone > have any experience with this type of protocols? > > Med vänlig hälsning / Best regards > > Sylvie > > @@@@@@@@@@@@@@@@@@@@@@@@ > Sylvie Le Guyader > Live Cell Imaging Unit > Dept of Biosciences and Nutrition > Karolinska Institutet > Sweden > office: +46 (0)8 608 9240 > LCI room: +46 (0)8 608 9248 > mobile: +46 (0) 73 733 5008 > >> -----Original Message----- >> From: Donnelly, Tom [mailto:[hidden email]] >> Sent: 24 September 2010 19:11 >> To: '[hidden email]' >> Subject: RE: water refill device for WI objective >> >> Sylvie; >> >> I have often thought the oil that Zeiss offers that has the same > refractive index of >> water might be a good solution here. Once the plate is coated ( probably > do that >> before starting scan) it would seem that little additional oil would be > needed. >> I have no experience with this and have no idea if it is compatible with > Nikon >> objectives, but just a thought. >> >> Tom >> >> Tom Donnelly >> Applied Precision, Inc. >> 1040 12th Ave. N.W. Issaquah, WA 98027-8929 >> (425)829-9414 (425)557-1055 fax >> [hidden email] >> http://www.appliedprecision.com >> >> >> >> -----Original Message----- >> From: Sylvie Le Guyader [mailto:[hidden email]] >> Sent: Friday, September 24, 2010 6:04 AM >> To: [hidden email] >> Subject: water refill device for WI objective >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all >> >> >> >> We have a 60x water objective NA1.2 by Nikon that we would like to use to > screen >> 96 and 384 well plates. Does anyone know of a water refill system that > would fit on >> this objective? Or has anyone tried the Olympus or the Leica water refill > device? Do >> they work well? Would any fit our objective? >> >> >> >> Thanks for your help! >> >> >> >> Med vänlig hälsning / Best regards >> >> >> >> Sylvie >> >> >> >> @@@@@@@@@@@@@@@@@@@@@@@@ >> >> Sylvie Le Guyader >> >> Live Cell Imaging Unit >> >> Dept of Biosciences and Nutrition >> >> Karolinska Institutet >> >> Sweden >> >> office: +46 (0)8 608 9240 >> >> mobile: +46 (0) 73 733 5008 >> >> >> >> >> This email message, together with any attachments, is for the sole use of > the >> intended recipient(s) and is the confidential information of Applied > Precision Inc. If >> you are not the intended recipient, your review, use, disclosure, copying > or >> dissemination of this email message or its attachments, or the information >> contained therein, is strictly prohibited. If you are not the intended > recipient or if you >> think this email was sent to you in error, please notify the sender by > reply email and >> delete this message and its attachments, as well as all copies, from your > system. > |
![]() ![]() |
Stephen Cody-2 |
![]() |
In reply to this post by Sylvie Le Guyader-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** G'day Sylvie, I've been building my own "lens re-watering" systems for water immersion objectives for about ten years. Back then I let most, if not all, the microscope companies know about it. They are very easy to build. My version uses a plastic pipette to add water via gravity feed. Any excess flow is sucked away via a peristaltic pump. All you need is some thin PE tubing (eg. Polyethylene Tubing I.D. = 0.58mm O.D. 0.96mm) Actual diameters not important, anything roughly those dimensions will do. You very slowly warm the tubing by holding it beside (NOT above), and about 5cm away from the flame of a Bunsen burner. Rotating slowly between thumbs and forefingers. Don't heat too quickly, slower the better. Keep the tubing under mild tension as you heat. As soon as it starts to stretch, remove from heat, BEFORE it breaks. That's the tricky part that needs practise. Don't throw away the failures, they are useful too, more later. Cut the PE tubing at it's thinnest point with a scalpel. You now have two plastic pipettes, that will not damage your lens. The failures end up curved, you can use these to suck away excess water (you'll have no shortage of these!) You now want to find a short length of some very large diameter Poly Vinyl tubing that will fit snugly around the top of your objective. eg. recently for a Leica WI lens we used approx. 30mm internal diameter tubing. Cut a thin slice of this tubing so that you have a ring, you can then fit this around the top of your objective, and have it just a little higher than the edges of the objective so that it forms a "moat", for those lenses that don't have a "moat". For lenses that do have a moat, allow the tubing to be just slightly proud of the moat. It should not be so large as to get in the way of the stage or your specimen. A trick taught to me by my colleague Robin Cole from the Ludwig: If the tubing is a little too large in diameter, you can shrink a small section by tightening a hose clamp around it (not when it is on an objective). Then heat the tubing (and clamp) with a heat gun (or very hot water might work too). Remove from heat, once cooled down to room temp., remove the hose clamp. Voila! Smaller tubing. Once fitted around the top of your objective, using an 18 Gauge hypodermic needle and a drilling motion with thumb and forefinger, drill two small holes through the very top of the tubing that now forms a moat. These holes will accommodate your re-watering pipette and suction tube. The pipette should be inserted so that it is just adjacent to the flat part of the objective that has the top lens. In other words you don't need it right next to the lens, a little way back so that it does not get banged against the coverslip. Capillary action will ensure the water will go between lens and coverslip. Sometimes you may want to use a little double sided tape to keep everything out of the way. Rig up a gravity feed with a 50ml syringe (without plunger) and 3-Way stop cock. I also use a 2ml syringe (with plunger) in the other port of the stop cock to prime the system, and for short term experiments eg, 2 hours you can use this to add water manually every 30 min. For longer term experiments eg. overnight, use gravity to do the work. The "failures"are bent and useful. Take one of these and insert through the other hole drilled into the moat, and connect to a peristaltic pump. Any excess flow onto the lens will spill into the moat, and as it builds up will be harmlessly sucked away via the pump into a beaker. In this way you don't need to carefully regulate flow rates. I have a PowerPoint describing this better and will send it via email to anyone who asks. I was going to write this up as a technique paper years ago, but the release of oils the same refractive index as water, and later commercial re-watering systems stopped me from doing so. I guess J. Microsc. may still be interested as a short Tech note. What do people think? Email me if you want the Re-watering gizmo How-to PowerPoint. We have successfully used this on time-lapse experiments at 37Deg C for three days continuous. Topping up the 50ml syringe each morning just to be sure. Cheers Steve Stephen H. Cody Imaging Research Fellow & Manager Monash Micro Imaging @ AMREP Monash University 6 Floor Burnet Tower Alfred Medical Research & Education Precinct 89 Commercial Rd, Melbourne, Australia, 3004 New email address: [hidden email] www.microimaging.monash.org Phone (Monash): (613) 990 30142 Phone (BakerIDI): (613) 8532 1580 On 24 September 2010 23:03, Sylvie Le Guyader <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all > > > > We have a 60x water objective NA1.2 by Nikon that we would like to use to > screen 96 and 384 well plates. Does anyone know of a water refill system > that would fit on this objective? Or has anyone tried the Olympus or the > Leica water refill device? Do they work well? Would any fit our objective? > > > > Thanks for your help! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Live Cell Imaging Unit > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Sweden > > office: +46 (0)8 608 9240 > > mobile: +46 (0) 73 733 5008 > > > -- Stephen H. Cody |
![]() ![]() |
Boswell, Carl A - (cboswell) |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I'd like to see your ppt file, Steve, if for no other reason than to admire your ingenuity and learn from the master. C Carl A. Boswell, Ph.D. Molecular and Cellular Biology University of Arizona 520-954-7053 FAX 520-621-3709 ----- Original Message ----- From: "Stephen Cody" <[hidden email]> To: <[hidden email]> Sent: Tuesday, September 28, 2010 8:00 PM Subject: Re: water refill device for WI objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** G'day Sylvie, I've been building my own "lens re-watering" systems for water immersion objectives for about ten years. Back then I let most, if not all, the microscope companies know about it. They are very easy to build. My version uses a plastic pipette to add water via gravity feed. Any excess flow is sucked away via a peristaltic pump. All you need is some thin PE tubing (eg. Polyethylene Tubing I.D. = 0.58mm O.D. 0.96mm) Actual diameters not important, anything roughly those dimensions will do. You very slowly warm the tubing by holding it beside (NOT above), and about 5cm away from the flame of a Bunsen burner. Rotating slowly between thumbs and forefingers. Don't heat too quickly, slower the better. Keep the tubing under mild tension as you heat. As soon as it starts to stretch, remove from heat, BEFORE it breaks. That's the tricky part that needs practise. Don't throw away the failures, they are useful too, more later. Cut the PE tubing at it's thinnest point with a scalpel. You now have two plastic pipettes, that will not damage your lens. The failures end up curved, you can use these to suck away excess water (you'll have no shortage of these!) You now want to find a short length of some very large diameter Poly Vinyl tubing that will fit snugly around the top of your objective. eg. recently for a Leica WI lens we used approx. 30mm internal diameter tubing. Cut a thin slice of this tubing so that you have a ring, you can then fit this around the top of your objective, and have it just a little higher than the edges of the objective so that it forms a "moat", for those lenses that don't have a "moat". For lenses that do have a moat, allow the tubing to be just slightly proud of the moat. It should not be so large as to get in the way of the stage or your specimen. A trick taught to me by my colleague Robin Cole from the Ludwig: If the tubing is a little too large in diameter, you can shrink a small section by tightening a hose clamp around it (not when it is on an objective). Then heat the tubing (and clamp) with a heat gun (or very hot water might work too). Remove from heat, once cooled down to room temp., remove the hose clamp. Voila! Smaller tubing. Once fitted around the top of your objective, using an 18 Gauge hypodermic needle and a drilling motion with thumb and forefinger, drill two small holes through the very top of the tubing that now forms a moat. These holes will accommodate your re-watering pipette and suction tube. The pipette should be inserted so that it is just adjacent to the flat part of the objective that has the top lens. In other words you don't need it right next to the lens, a little way back so that it does not get banged against the coverslip. Capillary action will ensure the water will go between lens and coverslip. Sometimes you may want to use a little double sided tape to keep everything out of the way. Rig up a gravity feed with a 50ml syringe (without plunger) and 3-Way stop cock. I also use a 2ml syringe (with plunger) in the other port of the stop cock to prime the system, and for short term experiments eg, 2 hours you can use this to add water manually every 30 min. For longer term experiments eg. overnight, use gravity to do the work. The "failures"are bent and useful. Take one of these and insert through the other hole drilled into the moat, and connect to a peristaltic pump. Any excess flow onto the lens will spill into the moat, and as it builds up will be harmlessly sucked away via the pump into a beaker. In this way you don't need to carefully regulate flow rates. I have a PowerPoint describing this better and will send it via email to anyone who asks. I was going to write this up as a technique paper years ago, but the release of oils the same refractive index as water, and later commercial re-watering systems stopped me from doing so. I guess J. Microsc. may still be interested as a short Tech note. What do people think? Email me if you want the Re-watering gizmo How-to PowerPoint. We have successfully used this on time-lapse experiments at 37Deg C for three days continuous. Topping up the 50ml syringe each morning just to be sure. Cheers Steve Stephen H. Cody Imaging Research Fellow & Manager Monash Micro Imaging @ AMREP Monash University 6 Floor Burnet Tower Alfred Medical Research & Education Precinct 89 Commercial Rd, Melbourne, Australia, 3004 New email address: [hidden email] www.microimaging.monash.org Phone (Monash): (613) 990 30142 Phone (BakerIDI): (613) 8532 1580 On 24 September 2010 23:03, Sylvie Le Guyader <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all > > > > We have a 60x water objective NA1.2 by Nikon that we would like to use to > screen 96 and 384 well plates. Does anyone know of a water refill system > that would fit on this objective? Or has anyone tried the Olympus or the > Leica water refill device? Do they work well? Would any fit our objective? > > > > Thanks for your help! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Live Cell Imaging Unit > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Sweden > > office: +46 (0)8 608 9240 > > mobile: +46 (0) 73 733 5008 > > > -- Stephen H. Cody |
![]() ![]() |
Neeraj Gohad-2 |
![]() |
In reply to this post by Stephen Cody-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Stephen, The system sounds very interesting, could you please send me the PPT file as well. You could surely write a tech note on this in either J. Microsc or even Microscopy Today. Thanks, Neeraj. Neeraj V. Gohad, Ph.D. Postdoctoral Fellow Okeanos Research Group Department of Biological Sciences 132 Long Hall Clemson University Clemson,SC-29634 Phone: 864-656-3597 Fax: 864-656-0435 Website: http://www.clemson.edu/okeanos -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Stephen Cody Sent: Tuesday, September 28, 2010 11:00 PM To: [hidden email] Subject: Re: water refill device for WI objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** G'day Sylvie, I've been building my own "lens re-watering" systems for water immersion objectives for about ten years. Back then I let most, if not all, the microscope companies know about it. They are very easy to build. My version uses a plastic pipette to add water via gravity feed. Any excess flow is sucked away via a peristaltic pump. All you need is some thin PE tubing (eg. Polyethylene Tubing I.D. = 0.58mm O.D. 0.96mm) Actual diameters not important, anything roughly those dimensions will do. You very slowly warm the tubing by holding it beside (NOT above), and about 5cm away from the flame of a Bunsen burner. Rotating slowly between thumbs and forefingers. Don't heat too quickly, slower the better. Keep the tubing under mild tension as you heat. As soon as it starts to stretch, remove from heat, BEFORE it breaks. That's the tricky part that needs practise. Don't throw away the failures, they are useful too, more later. Cut the PE tubing at it's thinnest point with a scalpel. You now have two plastic pipettes, that will not damage your lens. The failures end up curved, you can use these to suck away excess water (you'll have no shortage of these!) You now want to find a short length of some very large diameter Poly Vinyl tubing that will fit snugly around the top of your objective. eg. recently for a Leica WI lens we used approx. 30mm internal diameter tubing. Cut a thin slice of this tubing so that you have a ring, you can then fit this around the top of your objective, and have it just a little higher than the edges of the objective so that it forms a "moat", for those lenses that don't have a "moat". For lenses that do have a moat, allow the tubing to be just slightly proud of the moat. It should not be so large as to get in the way of the stage or your specimen. A trick taught to me by my colleague Robin Cole from the Ludwig: If the tubing is a little too large in diameter, you can shrink a small section by tightening a hose clamp around it (not when it is on an objective). Then heat the tubing (and clamp) with a heat gun (or very hot water might work too). Remove from heat, once cooled down to room temp., remove the hose clamp. Voila! Smaller tubing. Once fitted around the top of your objective, using an 18 Gauge hypodermic needle and a drilling motion with thumb and forefinger, drill two small holes through the very top of the tubing that now forms a moat. These holes will accommodate your re-watering pipette and suction tube. The pipette should be inserted so that it is just adjacent to the flat part of the objective that has the top lens. In other words you don't need it right next to the lens, a little way back so that it does not get banged against the coverslip. Capillary action will ensure the water will go between lens and coverslip. Sometimes you may want to use a little double sided tape to keep everything out of the way. Rig up a gravity feed with a 50ml syringe (without plunger) and 3-Way stop cock. I also use a 2ml syringe (with plunger) in the other port of the stop cock to prime the system, and for short term experiments eg, 2 hours you can use this to add water manually every 30 min. For longer term experiments eg. overnight, use gravity to do the work. The "failures"are bent and useful. Take one of these and insert through the other hole drilled into the moat, and connect to a peristaltic pump. Any excess flow onto the lens will spill into the moat, and as it builds up will be harmlessly sucked away via the pump into a beaker. In this way you don't need to carefully regulate flow rates. I have a PowerPoint describing this better and will send it via email to anyone who asks. I was going to write this up as a technique paper years ago, but the release of oils the same refractive index as water, and later commercial re-watering systems stopped me from doing so. I guess J. Microsc. may still be interested as a short Tech note. What do people think? Email me if you want the Re-watering gizmo How-to PowerPoint. We have successfully used this on time-lapse experiments at 37Deg C for three days continuous. Topping up the 50ml syringe each morning just to be sure. Cheers Steve Stephen H. Cody Imaging Research Fellow & Manager Monash Micro Imaging @ AMREP Monash University 6 Floor Burnet Tower Alfred Medical Research & Education Precinct 89 Commercial Rd, Melbourne, Australia, 3004 New email address: [hidden email] www.microimaging.monash.org Phone (Monash): (613) 990 30142 Phone (BakerIDI): (613) 8532 1580 On 24 September 2010 23:03, Sylvie Le Guyader <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all > > > > We have a 60x water objective NA1.2 by Nikon that we would like to use to > screen 96 and 384 well plates. Does anyone know of a water refill system > that would fit on this objective? Or has anyone tried the Olympus or the > Leica water refill device? Do they work well? Would any fit our objective? > > > > Thanks for your help! > > > > Med vänlig hälsning / Best regards > > > > Sylvie > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > Sylvie Le Guyader > > Live Cell Imaging Unit > > Dept of Biosciences and Nutrition > > Karolinska Institutet > > Sweden > > office: +46 (0)8 608 9240 > > mobile: +46 (0) 73 733 5008 > > > -- Stephen H. Cody |
![]() ![]() |
Sylvie Le Guyader-2 |
![]() |
In reply to this post by Boswell, Carl A - (cboswell)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Carl Sounds really good! I'd love to get that PP presentation if you do not mind! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet 14183 Huddinge Sweden office: +46 (0)8 608 9240 LCI room: +46 (0)8 608 9248 mobile: +46 (0) 73 733 5008 > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Carl Boswell > Sent: 29 September 2010 19:14 > To: [hidden email] > Subject: Re: water refill device for WI objective > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I'd like to see your ppt file, Steve, if for no other reason than to > your ingenuity and learn from the master. > C > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 > ----- Original Message ----- > From: "Stephen Cody" <[hidden email]> > To: <[hidden email]> > Sent: Tuesday, September 28, 2010 8:00 PM > Subject: Re: water refill device for WI objective > > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > G'day Sylvie, > > I've been building my own "lens re-watering" systems for water > immersion objectives for about ten years. Back then I let most, if not > all, the microscope companies know about it. They are very easy to > build. > > My version uses a plastic pipette to add water via gravity feed. Any > excess flow is sucked away via a peristaltic pump. > > All you need is some thin PE tubing (eg. Polyethylene Tubing I.D. = > 0.58mm O.D. 0.96mm) > Actual diameters not important, anything roughly those dimensions will do. > > You very slowly warm the tubing by holding it beside (NOT above), and > about 5cm away from the flame of a Bunsen burner. Rotating slowly > between thumbs and forefingers. Don't heat too quickly, slower the > better. > > Keep the tubing under mild tension as you heat. > > As soon as it starts to stretch, remove from heat, BEFORE it breaks. > That's the tricky part that needs practise. > > Don't throw away the failures, they are useful too, more later. > > Cut the PE tubing at it's thinnest point with a scalpel. You now have > two plastic pipettes, that will not damage your lens. > The failures end up curved, you can use these to suck away excess > water (you'll have no shortage of these!) > > You now want to find a short length of some very large diameter Poly > Vinyl tubing that will fit snugly around the top of your objective. > eg. recently for a Leica WI lens we used approx. 30mm internal > diameter tubing. Cut a thin slice of this tubing so that you have a > ring, you can then fit this around the top of your objective, and have > it just a little higher than the edges of the objective so that it > forms a "moat", for those lenses that don't have a "moat". For lenses > that do have a moat, allow the tubing to be just slightly proud of the > moat. It should not be so large as to get in the way of the stage or > your specimen. > > A trick taught to me by my colleague Robin Cole from the Ludwig: If > the tubing is a little too large in diameter, you can shrink a small > section by tightening a hose clamp around it (not when it is on an > objective). Then heat the tubing (and clamp) with a heat gun (or very > hot water might work too). Remove from heat, once cooled down to room > temp., remove the hose clamp. Voila! Smaller tubing. > > Once fitted around the top of your objective, using an 18 Gauge > hypodermic needle and a drilling motion with thumb and forefinger, > drill two small holes through the very top of the tubing that now > forms a moat. These holes will accommodate your re-watering pipette > and suction tube. The pipette should be inserted so that it is just > adjacent to the flat part of the objective that has the top lens. In > other words you don't need it right next to the lens, a little way > back so that it does not get banged against the coverslip. Capillary > action will ensure the water will go between lens and coverslip. > Sometimes you may want to use a little double sided tape to keep > everything out of the way. > > Rig up a gravity feed with a 50ml syringe (without plunger) and 3-Way > stop cock. I also use a 2ml syringe (with plunger) in the other port > of the stop cock to prime the system, and for short term experiments > eg, 2 hours you can use this to add water manually every 30 min. For > longer term experiments eg. overnight, use gravity to do the work. > > The "failures"are bent and useful. Take one of these and insert > through the other hole drilled into the moat, and connect to a > peristaltic pump. Any excess flow onto the lens will spill into the > moat, and as it builds up will be harmlessly sucked away via the pump > into a beaker. In this way you don't need to carefully regulate flow > rates. > > I have a PowerPoint describing this better and will send it via email > to anyone who asks. > > I was going to write this up as a technique paper years ago, but the > release of oils the same refractive index as water, and later > commercial re-watering systems stopped me from doing so. > > I guess J. Microsc. may still be interested as a short Tech note. What > do people think? Email me if you want the Re-watering gizmo How-to > PowerPoint. > > We have successfully used this on time-lapse experiments at 37Deg C > for three days continuous. Topping up the 50ml syringe each morning > just to be sure. > > Cheers Steve > > Stephen H. Cody > Imaging Research Fellow & Manager > Monash Micro Imaging @ AMREP > Monash University > 6 Floor Burnet Tower > Alfred Medical Research & Education Precinct > 89 Commercial Rd, Melbourne, Australia, 3004 > > New email address: > [hidden email] > > www.microimaging.monash.org > Phone (Monash): (613) 990 30142 > Phone (BakerIDI): (613) 8532 1580 > > > > > On 24 September 2010 23:03, Sylvie Le Guyader <[hidden email]> > wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear all > > > > > > > > We have a 60x water objective NA1.2 by Nikon that we would like to use > > screen 96 and 384 well plates. Does anyone know of a water refill system > > that would fit on this objective? Or has anyone tried the Olympus or the > > Leica water refill device? Do they work well? Would any fit our objective? > > > > > > > > Thanks for your help! > > > > > > > > Med vänlig hälsning / Best regards > > > > > > > > Sylvie > > > > > > > > @@@@@@@@@@@@@@@@@@@@@@@@ > > > > Sylvie Le Guyader > > > > Live Cell Imaging Unit > > > > Dept of Biosciences and Nutrition > > > > Karolinska Institutet > > > > Sweden > > > > office: +46 (0)8 608 9240 > > > > mobile: +46 (0) 73 733 5008 > > > > > > > > > > -- > Stephen H. Cody |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Interested parties I just received information from a company that makes a device for this application. They were not responding to this thread but after a review of their web site I saw that they have addressed this problem. I have never used it but I did look into the technology. Apparently it uses a new kind of micro-pump that uses Piezo deflection of a membrane to induce nanoflow. It looks good and I am waiting to get a demo of this in my lab. It looks to me that this is the company that makes a water replenishment system for Leica. They might have a device for other manufacturers objectives but I don't know. I am still looking into it. Further info: http://www.bartels-mikrotechnik.de/images/stories/components/download/Leica_Microsystems_Water_Immersion_Micro_Dispenser.pdf Dan On Oct 1, 2010, at 9:37 AM, Sylvie Le Guyader wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Carl Sounds really good! I'd love to get that PP presentation if you do not mind! Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader Live Cell Imaging Unit Dept of Biosciences and Nutrition Karolinska Institutet 14183 Huddinge Sweden office: +46 (0)8 608 9240 LCI room: +46 (0)8 608 9248 mobile: +46 (0) 73 733 5008 > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Carl Boswell > Sent: 29 September 2010 19:14 > To: [hidden email] > Subject: Re: water refill device for WI objective > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I'd like to see your ppt file, Steve, if for no other reason than to > your ingenuity and learn from the master. > C > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > University of Arizona > 520-954-7053 > FAX 520-621-3709 > ----- Original Message ----- > From: "Stephen Cody" <[hidden email]> > To: <[hidden email]> > Sent: Tuesday, September 28, 2010 8:00 PM > Subject: Re: water refill device for WI objective > > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > G'day Sylvie, > > I've been building my own "lens re-watering" systems for water > immersion objectives for about ten years. Back then I let most, if not > all, the microscope companies know about it. They are very easy to > build. > > My version uses a plastic pipette to add water via gravity feed. Any > excess flow is sucked away via a peristaltic pump. > > All you need is some thin PE tubing (eg. Polyethylene Tubing I.D. = > 0.58mm O.D. 0.96mm) > Actual diameters not important, anything roughly those dimensions will do. > > You very slowly warm the tubing by holding it beside (NOT above), and > about 5cm away from the flame of a Bunsen burner. Rotating slowly > between thumbs and forefingers. Don't heat too quickly, slower the > better. > > Keep the tubing under mild tension as you heat. > > As soon as it starts to stretch, remove from heat, BEFORE it breaks. > That's the tricky part that needs practise. > > Don't throw away the failures, they are useful too, more later. > > Cut the PE tubing at it's thinnest point with a scalpel. You now have > two plastic pipettes, that will not damage your lens. > The failures end up curved, you can use these to suck away excess > water (you'll have no shortage of these!) > > You now want to find a short length of some very large diameter Poly > Vinyl tubing that will fit snugly around the top of your objective. > eg. recently for a Leica WI lens we used approx. 30mm internal > diameter tubing. Cut a thin slice of this tubing so that you have a > ring, you can then fit this around the top of your objective, and have > it just a little higher than the edges of the objective so that it > forms a "moat", for those lenses that don't have a "moat". For lenses > that do have a moat, allow the tubing to be just slightly proud of the > moat. It should not be so large as to get in the way of the stage or > your specimen. > > A trick taught to me by my colleague Robin Cole from the Ludwig: If > the tubing is a little too large in diameter, you can shrink a small > section by tightening a hose clamp around it (not when it is on an > objective). Then heat the tubing (and clamp) with a heat gun (or very > hot water might work too). Remove from heat, once cooled down to room > temp., remove the hose clamp. Voila! Smaller tubing. > > Once fitted around the top of your objective, using an 18 Gauge > hypodermic needle and a drilling motion with thumb and forefinger, > drill two small holes through the very top of the tubing that now > forms a moat. These holes will accommodate your re-watering pipette > and suction tube. The pipette should be inserted so that it is just > adjacent to the flat part of the objective that has the top lens. In > other words you don't need it right next to the lens, a little way > back so that it does not get banged against the coverslip. Capillary > action will ensure the water will go between lens and coverslip. > Sometimes you may want to use a little double sided tape to keep > everything out of the way. > > Rig up a gravity feed with a 50ml syringe (without plunger) and 3-Way > stop cock. I also use a 2ml syringe (with plunger) in the other port > of the stop cock to prime the system, and for short term experiments > eg, 2 hours you can use this to add water manually every 30 min. For > longer term experiments eg. overnight, use gravity to do the work. > > The "failures"are bent and useful. Take one of these and insert > through the other hole drilled into the moat, and connect to a > peristaltic pump. Any excess flow onto the lens will spill into the > moat, and as it builds up will be harmlessly sucked away via the pump > into a beaker. In this way you don't need to carefully regulate flow > rates. > > I have a PowerPoint describing this better and will send it via email > to anyone who asks. > > I was going to write this up as a technique paper years ago, but the > release of oils the same refractive index as water, and later > commercial re-watering systems stopped me from doing so. > > I guess J. Microsc. may still be interested as a short Tech note. What > do people think? Email me if you want the Re-watering gizmo How-to > PowerPoint. > > We have successfully used this on time-lapse experiments at 37Deg C > for three days continuous. Topping up the 50ml syringe each morning > just to be sure. > > Cheers Steve > > Stephen H. Cody > Imaging Research Fellow & Manager > Monash Micro Imaging @ AMREP > Monash University > 6 Floor Burnet Tower > Alfred Medical Research & Education Precinct > 89 Commercial Rd, Melbourne, Australia, 3004 > > New email address: > [hidden email] > > www.microimaging.monash.org > Phone (Monash): (613) 990 30142 > Phone (BakerIDI): (613) 8532 1580 > > > > > On 24 September 2010 23:03, Sylvie Le Guyader <[hidden email]> > wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all >> >> >> >> We have a 60x water objective NA1.2 by Nikon that we would like to use >> screen 96 and 384 well plates. Does anyone know of a water refill system >> that would fit on this objective? Or has anyone tried the Olympus or the >> Leica water refill device? Do they work well? Would any fit our objective? >> >> >> >> Thanks for your help! >> >> >> >> Med vänlig hälsning / Best regards >> >> >> >> Sylvie >> >> >> >> @@@@@@@@@@@@@@@@@@@@@@@@ >> >> Sylvie Le Guyader >> >> Live Cell Imaging Unit >> >> Dept of Biosciences and Nutrition >> >> Karolinska Institutet >> >> Sweden >> >> office: +46 (0)8 608 9240 >> >> mobile: +46 (0) 73 733 5008 >> >> >> > > > > -- > Stephen H. Cody Dan Focht Bioptechs, Inc. 3560 Beck Rd. Butler, PA 16002 www.bioptechs.com P: (724)282-7145 F: (724)282-0745 [hidden email] |
![]() ![]() |
Ian Dobbie |
![]() |
In reply to this post by Stephen Cody-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Stephen Cody <[hidden email]> writes: > I've been building my own "lens re-watering" systems for water > immersion objectives for about ten years. Back then I let most, if not > all, the microscope companies know about it. They are very easy to > build. [snip detailed description] > I was going to write this up as a technique paper years ago, but the > release of oils the same refractive index as water, and later > commercial re-watering systems stopped me from doing so. > > I guess J. Microsc. may still be interested as a short Tech note. What > do people think? Email me if you want the Re-watering gizmo How-to > PowerPoint. I think this would be a very useful paper to have available. I think it would be well worth your time (and great for the rest of us!) doing a quick write up with a few diagrams/pictures to make clear exactly how it works. Thanks, Ian |
![]() ![]() |
Csúcs Gábor |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I'd also highly appreciate it, Gabor > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Stephen Cody<[hidden email]> writes: > >> I've been building my own "lens re-watering" systems for water >> immersion objectives for about ten years. Back then I let most, if not >> all, the microscope companies know about it. They are very easy to >> build. > [snip detailed description] > >> I was going to write this up as a technique paper years ago, but the >> release of oils the same refractive index as water, and later >> commercial re-watering systems stopped me from doing so. >> >> I guess J. Microsc. may still be interested as a short Tech note. What >> do people think? Email me if you want the Re-watering gizmo How-to >> PowerPoint. > I think this would be a very useful paper to have available. I think it > would be well worth your time (and great for the rest of us!) doing a > quick write up with a few diagrams/pictures to make clear exactly how it > works. > > Thanks, > > Ian -- Gabor Csucs Light Microscopy Centre, ETH Zurich Schafmattstrasse 18, HPM F16 CH-8093, Zurich, Switzerland Web: www.lmc.ethz.ch Phone: +41 44 633 6221 Mobile: +41 79 758 21 58 Fax: +41 44 632 1298 e-mail: [hidden email] |